An automated and standardized neural index to quantify patient-ventilator interaction
Critical Care volume 17, Article number: R239 (2013)
Abstract
Introduction
The aim of this study was to validate an automated, objective and standardized algorithm for quantifying and displaying patient-ventilator interaction.
Methods
Using a new method to detect patient-ventilator synchrony, the present study re-analyzed previously acquired and published data from 24 mechanically ventilated adult patients (Colombo et al., Crit Care Med. 2011 Nov;39(11):2452–7). Patient-ventilator interactions were evaluated by comparing ventilator pressure and diaphragm electrical activity (EAdi) waveforms, recorded during pressure support ventilation. The EAdi and ventilator pressure waveforms were analyzed for their timings (manually and automatically determined), and the error between the two waveforms was quantified. A new index of patient-ventilator interaction (NeuroSync index), which is standardized and automated, was validated and compared to manual analysis and previously published indices of asynchrony.
Results
The comparison of manual and automated detection methods produced high test-retest and inter-rater reliability (Intraclass correlation coefficient = 0.95). The NeuroSync index increased the sensitivity of detecting dyssynchronies, compared to previously published indices, which were found to only detect asynchronies.
Conclusion
The present study introduces an automated method and the NeuroSync index to determine patient-ventilator interaction with a more sensitive analysis method than those previously described. A dashboard-style of graphical display allows a rapid overview of patient-ventilator interaction and breathing pattern at the bedside.
Introduction
Severe patient-ventilator asynchrony, judged by visual inspection of airway pressure and flow waveforms, can be as high as 25% in intubated and spontaneously breathing patients with acute respiratory failure, and is associated with prolonged time on mechanical ventilation and increased use of sedation[1–3]. Recently, it was shown that this pneumatic waveform analysis considerably underestimates the prevalence of asynchronies and may even fail to reveal whether or not the patient is breathing[4].
The diaphragm electrical activity (EAdi) waveform is a reliable signal to monitor the patient’s neural respiratory drive[5] as well as patient-ventilator interaction[6]. The present study aimed to introduce and to test a new objective method of automatically detecting, quantifying, and displaying patient-ventilator interaction based on the measurements of EAdi and airway pressure waveforms.
Materials and methods
Data
The datasets used in the present study (43 datasets in total) are from 24 adult patients with acute respiratory failure of varying etiology, intubated and on pressure support ventilation, and were obtained from previously published material from Colombo et al.[4]. Each patient had a nasogastric tube with a multiple array of sensors for measuring EAdi (NAVA catheter, Maquet, Solna, Sweden). The EAdi was recorded in conjunction with the ventilator pressure (PV) and flow waveforms during 5-minute periods. All patients were receiving mechanical ventilation with a Servoi ventilator (Maquet).
EAdi signal processing is achieved by the Edi module in the Servoi ventilator, and ensures that changes in diaphragm position along the array are accounted for[5, 7] and that cardiac electrical activity is detected and replaced[5, 7].
As described by Colombo et al.[4], waveforms of ventilator pressure and EAdi were acquired from the RS232 interface at a sampling of 100 Hz, and recorded by means of dedicated software (NAVA tracker, Maquet).
General description of the new method to manually or automatically detect and quantify patient-ventilator interaction
Patient-ventilator interaction was evaluated by comparing the ventilator pressure and EAdi waveforms. The EAdi and ventilator pressure waveforms were analyzed with both manual and automated algorithms that detect their timings, and quantify the error between them. A new index of patient-ventilator interaction (NeuroSync index), which is standardized and automated (NeuroSyncAUTO), was compared to manual analysis (NeuroSyncMANU) and previously published indices of asynchrony.
Manual analysis: detection of neural (EAdi) and ventilator time points
Two experts manually analyzed all datasets twice with a visual display of the EAdi, PV, and flow waveforms. Cursors were placed at the onset of EAdi (EAdiON) and at about 1/3 of a decrease in EAdi from its peak (EAdiOFF). This cursor placement was based on data from a separate group of patients undergoing a spontaneous breathing trial with t-piece, showing that the onset of expiratory flow coincides with a decrease in peak EAdi by about 30% (unpublished observations).
Cursors were also placed, for each breath, at the onset of PV (beginning of ventilator pressurization) and end of PV.
Automatic analysis
Detection of neural (EAdi) and ventilator time points
Automatic detection of the onset of EAdi (beginning of neural inspiration) was obtained by detecting increases in EAdi, starting from the nadir (lowest point) of the EAdi. When a preset increase in EAdi (the EAdi threshold level) was reached, the time at the nadir was stored (onset of EAdi, EAdiON; Figure 1A, long-dashed vertical line). Three threshold levels were tested: 0.25, 0.5 and 1.0 μV. The amplitude at EAdiON was also stored. The EAdiOFF was automatically detected by finding when the EAdi had decreased to 70% of its peak (the EAdi termination level), and this was stored as the end of EAdi (EAdiOFF; Figure 1A, short dashed vertical lines).
The onset of pressure support (PSON) was automatically detected by searching for an increase in PV >3 cm H2O; when reached the time value obtained at the beginning of ventilator-delivered pressurization was stored as PSON (Figure 1B, examples 1-6). The termination of pressure support (PSOFF; Figure 1B, examples 1-6) was automatically detected by searching for the decrease in PV.
Defragmentation
To evaluate the influence of subventilatory efforts, data were processed both with and without a defragmentation method, that is, ignoring EAdi-triggered breaths of less than 0.15 μV*s (minimal area under the EAdi curve required to be called an effort) and pressure-detected breaths of less than 1.5 cm H2O*s (minimal area under PV curve required to be called an assist).
Description of neural index to evaluate patient-ventilator interaction
EAdi and PV timings were used to calculate an index (NeuroSync index) from both the manual (NeuroSyncMANU) and automated (NeuroSyncAUTO) detection methods.
Neural inspiratory detection periods were defined as segments from one detected EAdiOFF to the next EAdiOFF. Neural expiratory detection periods were defined as segments from one EAdiON to the next EAdiON (Figure 1A).
Each neural inspiratory detection period was divided into early and late segments using the EAdiON as the divider (Figure 1A). Depending on where PSON occurred, it was expressed as a percentage of either the early or late segment of the neural inspiratory detection period (Figure 1B). Thus, an early trigger error could range between −100% and 0% (a negative trigger error) and a late trigger error could range between 0% and +100% (positive trigger error).
In the same fashion, each neural expiratory detection period was divided into early and late segments using the EAdiOFF as the divider (Figure 1A). Depending on where PSOFF occurred, it was expressed as a percentage of either the early or late segment of the neural expiratory detection period (Figure 1B). Thus, an early cycle-off error could range between −100% and 0% and a late cycle-off error could range between 0% and 100%.
Events where EAdi and PV are completely dissociated are assigned 100% (Figure 1D).
A graphical presentation of the NeuroSync index is shown in Figure 1C. The figure depicts the neural inspiratory (y-axis) and expiratory detection (x-axis) periods and also has a box to indicate limits between synchrony (neural efforts matched to assist delivery are inside the box), dyssynchrony (neural efforts poorly related to assist delivery are outside the box), and asynchrony (neural efforts not related to assist delivery or vice versa). The limits of the box were set to ±33% difference between EAdiON and PSON as well as EAdiOFF and PSOFF, respectively.
The NeuroSync index was calculated by averaging the absolute values of the errors for all events.
Examples of synchrony, dyssynchrony, and asynchrony are described in Figure 1.
Comparison of indices
The NeuroSyncMANU and NeuroSyncAUTO indices were compared to the asynchrony index previously published by Colombo et al. (AIColombo)[4]. In that paper, three examiners with specific expertise in patient-ventilator interaction used the EAdi signal to verify the accuracy of flow-pressure waveform analysis described by Thille (AIThille)[1]. The asynchrony definitions used by Thille et al.[1] are presented in Table 1.
Neural (breathing) frequency (FN) was calculated from the EAdi signal. Ventilator frequency (FV) was calculated from PV.
Statistics
The intraclass correlation coefficient (ICC) was used for test-retest and inter-rater reliability (SPSS 16.0 for Windows (SPSS Inc, Chicago, IL, USA). Linear regression analysis was used to determine regression coefficients, intercepts, and determination coefficients. Unpaired comparisons were made using the Mann-Whitney rank sum test.
Results
Reliability of automated analysis
For the analysis of the datasets, the two expert analysts manually detected, on average, 4,562 (range 4,439 to 4,686) events (EAdi or PV events). ICCs for the NeuroSyncMANU index obtained by the two expert analysts during two repeated analyses agreed well and are presented in Table 2.
Figure 2A demonstrates the relationship between AIColombo versus NeuroSyncMANU for all events and Figure 2, Panel B displays only the asynchrony events. Figure 2C shows the relationship between NeuroSyncMANU and NeuroSyncAUTO with 0.5-μV trigger threshold. Table 3 provides ICCs between the NeuroSyncMANU and NeuroSyncAUTO indices at different trigger levels, with and without defragmentation.
Graphical display
Figure 3 provides dashboard-style graphs, displaying patient-ventilator interaction and breathing pattern in three different patients. Figure 3A demonstrates an example of good patient-ventilator interaction. Figure 3B and 3C depicts examples of poor patient-ventilator interaction.
Subventilatory efforts
Of the analyzed datasets, 61% had ≤2 subventilatory EAdi efforts per minute and only 6% had >8 subventilatory EAdi efforts per minute. The EAdi was higher (P = 0.019) for datasets that had ≤2 subventilatory EAdi efforts per minute (5.7 μV, n = 26) compared to those with >2 subventilatory EAdi efforts per minute (3.0 μV, n = 17).
With respect to breathing frequency, as depicted in Table 4, the ICCs between FN with automated and manual analysis were poor for breaths with EAdi amplitudes lower than 1 μV and excellent above 2 μV. Comparison of FV between automated and manual analysis for defragmented breaths resulted in an ICC of 1.0.
Discussion
The present study introduces a new method for automated quantification and graphical presentation of patient-ventilator interaction and breathing pattern, using the EAdi waveform as the reference. The NeuroSync index provides better distinction for the degrees of patient-ventilator interaction by classifying events falling inside the box as synchrony, outside the box as dyssynchrony, and the remainder as asynchrony. Strong ICCs for test-retest, inter-rater, and inter-method reliability suggest that the NeuroSync index and automated detection method are both valid and reliable[9].
The NeuroSync index - in combination with the graphical illustration - allows an understanding of the relative differences in timing between neural efforts and assist-delivery. The index, therefore, becomes insensitive to variances in breathing pattern which can occur with age and disease. For example, a trigger delay of 100 ms in a newborn having a neural inspiratory time of 300 ms results in a 33% error, and could be considered unacceptable. However, in an adult patient with a neural inspiratory time of 800 ms, the same trigger delay represents a 12% error and could be considered acceptable. The 33% error limits may be considered big, however, delays of this magnitude are empirically common in conventional pneumatically controlled modes[8].
A neural inspiratory effort modulates motor-unit firing rate and recruitment of the diaphragm, whose temporo-spatial summation yields the EAdi[5]. Hence, the EAdi signal if acquired and processed accurately represents the neural inspiratory drive to the diaphragm. The present study uses a recommended and standardized method to process EAdi[5]. Yet, the EAdi can be disturbed by other signals such as the electrocardiograph (ECG), thus impairing accurate determination of the onset and/or end of a neural effort. For example, an ECG occurring during the onset of the neural inspiration (would be detected and replaced by its previous value) and could give a negative trigger error (airway pressure increase before EAdi increase).
The concept of a negative trigger error deserves some consideration. A negative trigger error is when the pressurization occurs prior to the onset of the EAdi signal, and can occur anywhere between the 0% and −100% error range. In our analysis, negative trigger errors that were less than −33% error (that is, 0% to −33%) were classified as synchrony. When could this possibly be observed? As mentioned above, the ECG replacement algorithm could interfere with detection of the onset of the EAdi. In the present EAdi analysis, we estimated the maximum error to determine the onset and end of a single EAdi effort to be equivalent to the duration of P- or QRS-waveforms. When averaged over hundreds of breaths, this error would become minute.
Negative trigger errors falling between the −33% and −100% error were classified as dyssynchrony. Although one could consider a significant negative trigger error as an asynchronous event, we quantified them and classified them as dyssynchrony in order to provide symmetry of our numerical and graphical representation. In the present study, these events were rare: only 0.9% of the total number of events fell into the category of negative trigger error and were classified as dyssynchrony (data not shown). Recent work by Akoumianaki et al.[9] in sedated and mechanically ventilated patients demonstrated respiratory entrainment, where patients’ neural efforts shortly follow ventilator pressurization. In theory, if this time delay would fall between −33 and −100% negative trigger error, it would be classified as dyssynchrony.
Naturally, extreme negative trigger errors are actually assist-without-Eadi and were classified as asynchrony, and given an error of 100%, as there is no patient interaction (no effort) associated with the ventilator.
Our analysis showed that subventilatory efforts are rare and typically related to very low EAdi amplitudes (<2 to 3 μV) and that their elimination has its greatest value at sensitive trigger/threshold levels (0.25 μV). A problem of subventilatory EAdi efforts is that if they fail to initiate assist, the event is classified as EAdi-without-assist (ineffective effort, also known as wasted effort), whereas if assist is initiated it is classified as assist-without-EAdi (auto-triggering). Subventilatory EAdi efforts introduced uncertainties in determining neural breathing pattern and with regards to the agreement between manual and automated analysis to determine FN, it was clear that low EAdi amplitude worsened the reliability. This underlines the importance of a good signal to noise ratio for automated analysis. It must be noted that very low or no EAdi signal that does not reach the EAdi threshold does not preclude that patients may be activating inspiratory muscles other than the diaphragm. The nature of these subventilatory efforts is unknown, but is most likely related to spontaneous firing or recruitment of diaphragm motor units. This is to be distinguished from signal disturbances, which are typically much higher in amplitude, and managed by other software algorithms.
Comparison of NeuroSync definitions to previous asynchrony definitions
The group of Thille et al.[1] was the first to describe and quantify major asynchronies, such as wasted efforts and auto-triggering, using only airway pressure and flow waveforms, albeit without the EAdi as a reference.
The NeuroSync event defined as EAdi-without-assist corresponds to ineffective triggering with the AIThille. An inspiratory effort not rewarded by a ventilator breath is a failure for a triggered mode and is the asynchrony predominantly associated with adverse patient outcomes[1, 2]. As ineffective triggering typically relates to a failure of the conventional ventilator flow and pressure sensors to detect an inspiratory effort, it is not surprising that the prevalence of ineffective triggering is greatly underestimated by airway flow and pressure detection[4].
The NeuroSync event defined as assist-without-EAdi resembles auto-triggering with AIThille. If not induced by backup modes during apnea, auto-triggering is another faulty condition where the ventilator triggers and cycles off uncontrollably, and hyperventilates the patient. Auto-triggering is a very difficult asynchrony to detect with AIThille, because there is no true patient reference to validate the ventilator triggering[10].
To describe dyssynchrony, the method of Thille et a l.[1] involved detection of short and prolonged cycles (Table 1). Considering the natural variability in breathing, however, the significance of detecting these remains unclear[11, 12]. The closest comparison to the NeuroSync index for short cycles would be late triggering and early cycling-off values, which would fall outside the box, that is, in the upper left quadrants in Figures 1 and3. Long cycles (Figure 1 and Table 1) are likely to be associated with early trigger (lower quadrants) and/or delayed cycling-off (right-side quadrants) or repeated EAdi during assist (asynchrony).
The AIThille index[1] also includes double-triggering, an event corresponding to multiple-assist-during-EAdi with the NeuroSync index. Multiple-assist-during-EAdi reflects repeated trigger and cycling-off errors during the same neural effort, which are classified as dyssynchrony. It should be noted that in assist-volume control, double triggering is a severe asynchrony associated with excessive tidal volumes[13]. In non-flow and volume-regulated modes, double triggering would only cause a timing error with a short interruption of the inspiratory assist during an inspiratory effort.
The NeuroSync index also introduces another type of asynchrony labeled multiple-EAdi-during-assist, a severe type of asynchrony where the ventilator is delivering one breath for several neural inspiratory efforts. The AIThille has no counterpart for multiple-EAdi-during-assist.
Because EAdi-without-assist, assist-without-EAdi, and multiple-EAdi-during-assist all describe failures of the ventilator trigger and cycling-off functions, these events were labeled as 100% trigger error and 100% cycling-off error, and called asynchrony.
In the context of the above discussion it is important to note that AIColombo[4] significantly increases the sensitivity to detect asynchrony compared to AIThille.
Our results show that both the AIThille and the EAdi-verified AIColombo[4] were not designed to detect timing errors between neural effort and assist-delivery (that is, they only detect asynchrony), whereas the NeuroSync index has added the ability to also quantify dyssynchrony and synchrony. This is evidenced in Figure 2A showing that the NeuroSync index reached 40% before AIColombo surpassed 10%. The close association between indices when AIColombo exceeds 10% (Figure 2B) shows that asynchronies are detected by both indices. As evidenced by a close relationship between the asynchronous events detected with the NeuroSync index and AIColombo, most asynchronies defined by the NeuroSync index provide information similar to that of the pressure- and flow-based asynchrony index described by Thille et al.[1].
Another index of asynchrony based on EAdi was described by Beck et al.[14], where the sum of trigger delays and cycling-off delays (determined manually) were expressed as a percentage of the total neural respiratory cycle[8, 14–16]. The NeuroSync index can be considered a development of the previously described EAdi-based index.
This new method has important clinical implications. The NeuroSync index provides real-time detection and quantification of: (i) patient ventilator asynchrony (of different types), (ii) dyssynchrony and (iii) synchrony. The method allows objective evaluation of patient neural breathing pattern and the ventilator performance, and could be used to adjust ventilator settings in order to optimize patient-ventilator interaction. Improved matching of patient and ventilator timings could enhance lung-distending pressure and ventilatory efficiency. Moreover, no study has yet shown any negative outcome of timing errors, related specifically to trigger and cycle-off delays (dyssynchrony). Asynchrony has been associated with prolonged time on mechanical ventilation[1, 2]. The NeuroSync index, therefore, could provide a tool for future studies to determine acceptable limits of dyssynchrony. It should be pointed out that since the EAdi signal is a pneumatically independent signal, it is not affected by leaks, implying that the analysis presented could be applied reliably in intubated infants with uncuffed endotracheal tubes and during non-invasive ventilation.
Conclusion
The present study introduces an automated method and the NeuroSync index to objectively determine patient-ventilator interaction with higher precision than previous methods. A dashboard style of graphical display allows a rapid overview of patient-ventilator interaction and breathing pattern.
Key messages
-
The diaphragm electrical activity is a useful signal for evaluating and monitoring patient-ventilator interaction.
-
The NeuroSync index (an automated and standardized index to quantify patient-ventilator interaction) was found to be reproducible and correlated to manual analysis by experts.
-
In mechanically ventilated adult patients, events can now be classified in an objective fashion as asynchronous, dyssynchronous or synchronous.
-
The NeuroSync index determines patient-ventilator interaction with more sensitive analysis than previous methods.
-
A dashboard style of graphical display allows a rapid overview and quantification of patient-ventilator interaction and breathing pattern at the bedside.
Abbreviations
- AI:
-
Colombo asynchrony index based on the definition of Colombo et al.[4]
- AIThille:
-
asynchrony index based on the definition of Thille et a l.[1]
- EAdi:
-
Electrical activity of the diaphragm
- EAdiOFF:
-
Manually or automatically determined 1/3 decrease in electrical activity of the diaphragm
- EAdiON:
-
Manually or automatically determined beginning of electrical activity of the diaphragm breath
- ECG:
-
Electrocardiograph
- FN:
-
Frequency of neural (patient) breaths
- FV:
-
Frequency of ventilator-delivered breaths
- ICC:
-
Intraclass correlation coefficient
- NeuroSync:
-
index index for patient-ventilator interaction based on automated algorithms described in the present paper
- indexAUTO:
-
Index for patient-ventilator interaction based on automated algorithms described in the present paper with automated selection of timings
- NeuroSyncMANU index:
-
Index for patient-ventilator interaction based on automated algorithms described in the present paper with manual selection of timings
- PEEP:
-
Positive end-expiratory pressure
- PSOFF:
-
End of pressure delivery
- PSON:
-
Start of pressure delivery
- PV:
-
Ventilator-delivered pressure.
References
Thille AW, Rodriguez P, Cabello B, Lellouche F, Brochard L: Patient-ventilator asynchrony during assisted mechanical ventilation. Intensive Care Med 2006, 32: 1515-1522. 10.1007/s00134-006-0301-8
de Wit M, Miller KB, Green DA, Ostman HE, Gennings C, Epstein SK: Ineffective triggering predicts increased duration of mechanical ventilation. Crit Care Med 2009, 37: 2740-2745. 10.1097/CCM.0b013e3181a98a05
de Wit M, Pedram S, Best AM, Epstein SK: Observational study of patient-ventilator asynchrony and relationship to sedation level. J Crit Care 2009, 24: 74-80. 10.1016/j.jcrc.2008.08.011
Colombo D, Cammarota G, Alemani M, Carenzo L, Barra FL, Vaschetto R, Slutsky AS, Della Corte F, Navalesi P: Efficacy of ventilator waveforms observation in detecting patient-ventilator asynchrony. Crit Care Med 2011, 39: 2452-2457. 10.1097/CCM.0b013e318225753c
Aldrich T, Sinderby C, McKenzie DK, Estenne M, Gandevia SC: ATS/ERS Statement on respiratory muscle testing. Am J Respir Crit Care Med 2001, 166: 518-624.
Parthasarathy S, Jubran A, Tobin MJ: Assessment of neural inspiratory time in ventilator-supported patients. Am J Respir Crit Care Med 2000, 162: 546-552. 10.1164/ajrccm.162.2.9901024
Sinderby CA, Beck JC, Lindström LH, Grassino AE: Enhancement of signal quality in esophageal recordings of diaphragm EMG. J Appl Physiol 1997, 82: 1370-1377.
Spahija J, de Marchie M, Albert M, Bellemare P, Delisle S, Beck J, Sinderby C: Patient-ventilator interaction during pressure support ventilation and neurally adjusted ventilatory assist. Crit Care Med 2010, 38: 518-526. 10.1097/CCM.0b013e3181cb0d7b
Akoumianaki E, Lyazidi A, Rey N, Matamis D, Perez-Martinez N, Giraud R, Mancebo J, Brochard L, Marie Richard JC: Mechanical ventilation-induced reverse triggered breaths: a frequently unrecognized form of neuro-mechanical coupling. Chest 2013, 143: 927-938. 10.1378/chest.12-1817
Sinderby C, Beck J: Proportional assist ventilation and neurally adjusted ventilatory assist–better approaches to patient ventilator synchrony? Clin Chest Med 2008, 29: 329-342. 10.1016/j.ccm.2008.01.007
Schmidt M, Demoule A, Cracco C, Gharbi A, Fiamma MN, Straus C, Duguet A, Gottfried SB, Similowski T: Neurally adjusted ventilatory assist increases respiratory variability and complexity in acute respiratory failure. Anesthesiology 2010, 112: 670-681. 10.1097/ALN.0b013e3181cea375
Colombo D, Cammarota G, Bergamaschi V, De Lucia M, Corte FD, Navalesi P: Physiologic response to varying levels of pressure support and neurally adjusted ventilatory assist in patients with acute respiratory failure. Intensive Care Med 2008, 34: 2010-2018. 10.1007/s00134-008-1208-3
Pohlman MC, McCallister KE, Schweickert WD, Pohlman AS, Nigos CP, Krishnan JA, Charbeneau JT, Gehlbach BK, Kress JP, Hall JB: Excessive tidal volume from breath stacking during lung-protective ventilation for acute lung injury. Crit Care Med 2008, 36: 3019-3023. 10.1097/CCM.0b013e31818b308b
Beck J, Tucci M, Emeriaud G, Lacroix J, Sinderby C: Prolonged neural expiratory time induced by mechanical ventilation in infants. Pediatr Res 2004, 55: 747-754. 10.1203/01.PDR.0000119368.21770.33
Moerer O, Beck J, Brander L, Costa R, Quintel M, Slutsky AS, Brunet F, Sinderby C: Subject-ventilator synchrony during neural versus pneumatically triggered non-invasive helmet ventilation. Intensive Care Med 2008, 34: 1615-1623. 10.1007/s00134-008-1163-z
Bordessoule A, Emeriaud G, Morneau S, Jouvet P, Beck J: Neurally Adjusted Ventilatory Assist (NAVA) improves patient-ventilator interaction in infants compared to conventional ventilation. Pediatr Res 2012, 72: 194-202. 10.1038/pr.2012.64
Acknowledgements
The research was funded in part by the RS McLaughlin Foundation.
Author information
Authors and Affiliations
Corresponding author
Additional information
Competing interests
Drs. Beck and Sinderby have made inventions related to neural control of mechanical ventilation that are patented. The license for these patents belongs to Maquet Critical Care. Future commercial uses of this technology may provide financial benefit to Drs. Beck and Sinderby through royalties. Dr Beck and Dr Sinderby each own 50% of Neurovent Research Inc (NVR). NVR is a research and development company that builds the equipment and catheters for research studies. NVR has a consulting agreement with Maquet Critical Care. A Slutsky is a paid consultant to Maquet Critical Care.
Authors’ contributions
CS conceived the study, described the algorithms for the analysis, and drafted the manuscript. SL performed data analysis and statistical analysis. DC collected the data, and revised the manuscript. GC collected the data and revised the manuscript. AS was involved with data interpretation and manuscript preparation. PN collected the data and revised the manuscript. JB participated in the design of the study and manuscript preparation. All authors read and approved the final manuscript.
Authors’ original submitted files for images
Below are the links to the authors’ original submitted files for images.
Rights and permissions
About this article
Cite this article
Sinderby, C., Liu, S., Colombo, D. et al. An automated and standardized neural index to quantify patient-ventilator interaction. Crit Care 17, R239 (2013). https://doi.org/10.1186/cc13063
Received:
Accepted:
Published:
DOI: https://doi.org/10.1186/cc13063