- Research
- Open Access
- Published:
Serial measurement of pancreatic stone protein for the early detection of sepsis in intensive care unit patients: a prospective multicentric study
Critical Care volume 25, Article number: 151 (2021)
Abstract
Background
The early recognition and management of sepsis improves outcomes. Biomarkers may help in identifying earlier sub-clinical signs of sepsis. We explored the potential of serial measurements of C-reactive protein (CRP), procalcitonin (PCT) and pancreatic stone protein (PSP) for the early recognition of sepsis in patients hospitalized in the intensive care unit (ICU).
Methods
This was a multicentric international prospective observational clinical study conducted in 14 ICUs in France, Switzerland, Italy, and the United Kingdom. Adult ICU patients at risk of nosocomial sepsis were included. A biomarker-blinded adjudication committee identified sepsis events and the days on which they began. The association of clinical sepsis diagnoses with the trajectories of PSP, CRP, and PCT in the 3 days preceding these diagnoses of sepsis were tested for markers of early sepsis detection. The performance of the biomarkers in sepsis diagnosis was assessed by receiver operating characteristic (ROC) analysis.
Results
Of the 243 patients included, 53 developed nosocomial sepsis after a median of 6 days (interquartile range, 3–8 days). Clinical sepsis diagnosis was associated with an increase in biomarkers value over the 3 days preceding this diagnosis [PSP (p = 0.003), PCT (p = 0.025) and CRP (p = 0.009)]. PSP started to increase 5 days before the clinical diagnosis of sepsis, PCT 3 and CRP 2 days, respectively. The area under the ROC curve at the time of clinical sepsis was similar for all markers (PSP, 0.75; CRP, 0.77; PCT, 0.75).
Conclusions
While the diagnostic accuracy of PSP, CRP and PCT for sepsis were similar in this cohort, serial PSP measurement demonstrated an increase of this marker the days preceding the onset of signs necessary to clinical diagnose sepsis. This observation justifies further evaluation of the potential clinical benefit of serial PSP measurement in the management of critically ill patients developing nosocomial sepsis.
Trial registration The study has been registered at ClinicalTrials.gov (no. NCT03474809), on March 16, 2018. https://www.clinicaltrials.gov/ct2/show/NCT03474809?term=NCT03474809&draw=2&rank=1.
Background
Sepsis, if not recognized and managed early, may evolve rapidly into life-threatening septic shock and multiple organ failure [1,2,3]. Sepsis and septic shock remain challenging global health problems associated with persistently high morbidity and mortality; in 2017, an estimated 48.9 million cases of sepsis were recorded worldwide, with 11.0 million related deaths, representing one-fifth of all causes of death [4]. Accordingly, guidelines systematically emphasize the early recognition and aggressive management of sepsis, with combined early antibiotic treatment and support to prevent organ dysfunction [1, 5, 6]. Currently, the confirmation of sepsis diagnosis is based largely on nonspecific clinical signs, laboratory findings and medical scores, which are usually obtained after sepsis onset. Despite extensive research, no biomarker has been identified with the capacity to detect sepsis quickly enough and with a high degree of diagnostic accuracy [7]. C-reactive protein (CRP) is a well characterized inflammatory marker widely used to help in the diagnosis of infection. Procalcitonin (PCT) has been extensively evaluated in the last 20 years as a marker of bacteremia. Even if CRP and PCT are commonly used in the context of the diagnosis of sepsis, both have shown suboptimal performance [8].
Pancreatic stone protein (PSP) is a C-type lectin protein that triggers polymorphonuclear cell activation and has shown proinflammatory activity in vitro [9]. In an unselected cohort of critically ill adults, PSP was found to be superior to PCT and other sepsis biomarkers for the accurate identification of sepsis [10]. An increase in PSP level preceding the development of sepsis has recently been demonstrated in a cohort of severely burnt patient [11]. The diagnostic performance of PSP, alone and in combination with other markers or clinical scores, was evaluated further in several studies conducted in adults, children, and neonates, in both, intensive care units (ICUs) and emergency departments [12]. In contrast to PCT and CRP, PSP was found to be a prognostic marker for ICU mortality [13,14,15]. In addition, point-of-care measurement of CRP and PCT is not common in ICU, while PSP can be measured with a ‘point-of-care’ device within 5 min using a single drop of whole blood [16], leading the way for simple, on-demand, around-the-clock, serial biomarker assessments instead of one-off measurements upon the clinical suspicion of sepsis.
We hypothesized that timely identification of changes in biomarker levels may help identify sepsis before the onset of clinical signs. We therefore designed a multicentric international prospective blind observational clinical study to explore the ability of serial PSP measurements to identify nosocomial sepsis before the onset of clinical signs required to clinically diagnose sepsis in an adult ICU population, and compare it with CRP and PCT.
Study design and methods
Study design
This multicentric biomarker-blinded prospective observational clinical study was conducted in 14 ICUs in France, Switzerland, Italy, and the United Kingdom, see Additional file 1: Table 1. Approval was obtained from the ethics committees of all participating sites. The study was registered at ClinicalTrials.gov (no. NCT03474809).
Patient population
All patients older than 18 years old admitted to the ICU and expected to stay at least 7 days and/or to be mechanically ventilated for at least 5 consecutive days were screened for study inclusion (Additional file 1:Table 2). Patients with a clinical diagnosis of sepsis or suspicion of sepsis at the admission were not included. Written informed consent to participation and any study-related assessment was obtained from all patients before their inclusion in the study. Written informed consent was obtained according to the specific requirements of each center’s ethics committee.
At least 40 patients who developed sepsis after study enrolment were planned to be included. Based on recently published data [17,18,19], the expected incidence of sepsis in the targeted ICU population was estimated conservatively to be 15%. This value corresponded to a sample of 267 patients, to which we added 10% to compensate for withdrawal and loss to follow-up.
Trial procedure and definitions
Patients were managed according to the centers’ standard clinical practices, including those for the diagnosis, assessment, and treatment of sepsis. Blood samples were collected daily for biomarker (PCT, CRP, and PSP) measurements. Patients were followed until death or discharge from the ICU or for 30 days, whichever occurred first. The Sepsis-3 criteria were used to define sepsis [1], and microbiological procedures were performed to diagnose infections responsible for sepsis according to local protocol.
Measurement of biomarkers
Daily PSP levels were determined at the end of the study period with the CE-marked IVD PSP capsule on the point-of-care abioSCOPE® device (Abionic SA; Additional file 2: Figure 1). The 5th and 95th percentiles of PSP values in healthy adults are 25.0 to 60.7 ng/ml (median: 41.7 ng/ml) (Abionic internal data). CRP and PCT measurements were performed by a central laboratory using a Tina-quant® C-Reactive Protein Gen.3 (Roche Diagnostics) and a Liaison® Brahms PCT® for PCT (DiaSorin), respectively. All instruments were used according to the manufacturers’ instructions. The upper limit of the normal range of the CRP assay is 5 mg/l, and of the PCT assay 0.1 ng/ml, as per the manufacturer’s information.
External adjudication committee
An independent endpoint adjudication committee (EAC) composed of three ICU experts was formed. Two non-chair EAC members retrospectively and independently reviewed case report forms for each patient and determined whether a septic event had occurred during the patient’s ICU stay (including on the day of inclusion), and on which day it started. Day 0 sepsis was determined using all information collected in the study’s eCRF and in accordance with the proposal published by Lambden et al. [20]. Discordant determinations were arbitrated by the EAC chair. The EAC had access to all case report forms, including results of all microbiological investigations, CRP and PCT measurements taken in the ICUs as part of standard care, but was blinded to centrally measured PSP, PCT, and CRP levels and the investigators’ diagnoses of sepsis.
Patient-related data
Data collected at the time of inclusion were: demographic characteristics, reason for ICU admission, medical history, and Charlson comorbidity index. Additional clinical data were collected on day 1 and daily thereafter when measurements were performed as part of standard care; these data included vital signs (temperature, heart rate, and blood pressure), Sequential Organ Failure Assessment (SOFA) score, PaO2/FiO2 ratio, hematological data, clinical chemistry parameters, need for organ support (ventilation and use of vasoactive drugs) and microbiological data. In addition to the collection of data on antibiotic treatments (drug name, dose, schedule, route, and indication), meticulous assessments for sepsis detection were carried out daily until ICU discharge, death, withdrawal, or day 30, whichever came first.
Statistical analyses
Discrete values are described as counts (percentages) and continuous variables are described as means with standard errors or medians with interquartile ranges (IQRs), as appropriate. The missing biomarker values were imputed by carrying the last observation forward, and multiple consecutive missing values were reported as missing. Based on EAC decisions, the patients were allocated to sepsis (a septic event occurred during the study period) or no-sepsis (no septic event during the study period) groups. In the sepsis group, the event day was set as the day on which the first septic events began, according to the EAC. In the no-sepsis group, the event day was set to day 7 of patients’ ICU stays, when more than 50% of adjudicated septic events occurred, or the days of discharge for patients with ICU stays of fewer than 7 days. Biomarkers average trajectories were visualized accordingly. All statistical analyses were performed using R version 4.0.2.
Biomarkers dynamics and progression toward sepsis
To explore the dynamics of serial biomarker measurement throughout the study period, we modeled biomarker courses in the 3 days prior to EAC sepsis diagnoses by fitting a linear mixed-effect model with PSP, CRP, and PCT as response and the following explanatory variables: patient-specific random effect, group, day-to-event and group by day-to-event interaction as fixed-effects (estimates are provided in Additional file 1: Table 3). According to Van Breukelen [21] “testing absence of group by time interaction is equivalent to testing the hypothesis of no group effect on the change” in a response variable over days.
Assessment of biomarker accuracy in diagnosing sepsis
The joint performance of the biomarkers in the diagnosis of sepsis on the day it occurred as per EAC decision was assessed using receiving operating characteristic (ROC) analysis. To account for differences among study sites and incidences of nosocomial sepsis, a linear mixed-effects logistic regression model was used with the center serving as a random intercept and biomarkers serving as fixed events to determine diagnostic accuracy, sensitivity, specificity, and the positive and negative likelihood ratios (estimates are provided in Additional file 1: Table 4) [22]. Cut-off values of PSP, CRP and PCT have been estimated without adjusting for center-effect.”
Results
Baseline characteristics of the study population
From June 2018 to March 2019, 297 patients were enrolled in this study. The mean patients recruited per site was 21, and the median number was 15 (IQR: 8 to 23) (Additional file 1: Table 1b). Fifty-four of these patients were excluded from the analysis due to the presence of infection and/or sepsis at the time of inclusion as adjudicated by the EAC (n = 33), ICU stays < 48 h (n = 15), withdrawal (n = 5), and screening failure (n = 1) (Fig. 1). The characteristics of the remaining 243 patients are summarized in Table 1. The cohort contained 153 (63%) men and had a median age of 65 years (IQR: 54 to 73) and a median SOFA score of 6 (IQR: 5 to 9) at the time of admission. The most common comorbidities were cancer (16%) and heart (15%), pulmonary (13%), renal (12%), and hepatic (7%) diseases. Vasopressors were administered to 39% of patients, and 77% of patients required mechanical ventilation. The main reasons for ICU admission were acute central nervous system diseases (39%), cardiovascular diseases (31%), and abdominal/digestive conditions (19%; Table 1). The median ICU length of stay of the entire cohort was 9 days (IQR: 6 to 17); 7 days (IQR: 5 to 13 days) for the “no-sepsis” patients and 21 days for patients who developed sepsis (IQR: 14 to 27 days). ICU mortality rate in the study cohort was 16%. The baseline clinical characteristics of patients who did and did not develop sepsis during the study period were similar.
Flow of patient enrollment. There were 4 (0.14%) missing values for PSP (0; 0.0% after imputation), 8 (0.35%) for CRP (1; 0.04% after imputation) and 8 (0.28%) for PCT (2; 0.07% after imputation). CRP C-reactive protein, EAC endpoint adjudication committee, ICU intensive care unit, PCT procalcitonin, PSP pancreatic stone protein
According to the EAC decisions, 21.8% (n = 53) of the patients developed sepsis and 78.2% (n = 190) did not. The EAC chair arbitrated disagreements about sepsis presence/absence and day of initiation in 31% of cases. Two thirds of them concerned the date of onset of the sepsis and were resolved by the chair of the committee. The median interval from study inclusion to sepsis development was 6 days (IQR: 3 to 8 days). The characteristics of patients who developed sepsis are provided in Table 2. Sepsis originated most frequently from respiratory tract infections (62%), followed by bloodstream infections (11%) and urinary tract (9%) infections.
Time courses of biomarkers for sepsis prediction
The time courses of biomarkers until sepsis (“sepsis” group), and until day 7 or discharge for ICU stays of fewer than 7 days (“no-sepsis” group) are shown in Fig. 2. All biomarker values increased for several days before the clinical diagnosis of sepsis (PSP, 5 days; PCT, 3 days; CRP, 2 days). Clinical sepsis diagnosis was associated with an increase in biomarkers over the 3 days preceding this diagnosis: PSP (p = 0.003), PCT (p = 0.025) and CRP (p = 0.009). Furthermore, estimates of exploratory variables are provided in the Additional file 1: Table 3.
Association of pancreatic stone protein, procalcitonin and C-reactive protein increase with sepsis. Daily mean values ± standard errors of the mean for the sepsis group are from the 6 days preceding the endpoint adjudication committee diagnosis of sepsis. For the no-sepsis group, day 0 was set as intensive care unit day 7 or the day of discharge for intensive care unit stays < 7 days. The p values are for the association of clinical sepsis diagnosis with a continuous biomarker increase in the preceding 3 days (pancreatic stone protein, p = 0.003; procalcitonin, p = 0.025; C-reactive protein, p = 0.009). ns are numbers of observations per day and group. PSP pancreatic stone protein, PCT procalcitonin, CRP C-reactive protein, EAC endpoint adjudication committee, ICU intensive care unit
Accuracy of individual biomarkers and their combination in sepsis diagnosis
Areas under receiver operating characteristic curves (AUROCCs) on the day of sepsis diagnosis according to the blinded EAC were similar for all three biomarkers {PSP, 0.75 [95% confidence interval (CI) 0.67–0.82]; CRP, 0.77 [95% CI 0.69–0.84] PCT, 0.75 [95% CI 0.68–0.83]} (Table 3 and Additional file 3: Figure 2). The combined use of PSP with CRP improved the accuracy to 0.79 (95% CI 0.72–0.86). The addition of PCT to this combination did not further improve the accuracy or sensitivity, and only marginally improved the specificity. Estimates for the mixed-effects models for which the diagnostic performances are reported in Table 3 are provided in the Additional file 1: Table 4. Estimated cut-off values of biomarkers are 290.5 ng/ml for PSP, 167.2 mg/l for CRP and 0.94 ng/ml for PCT.
Discussion
This prospective multicentric study evaluated the potential of serial host protein biomarkers measurement, in particular the relatively new marker PSP, for the early identification of nosocomial sepsis in ICU patients. In the study cohort of critically ill patients at high risk of complications, in which one in five developed sepsis, the baseline clinical characteristics, including severity scores and the need for mechanical ventilation and vasopressor support, were similar among patients who did and did not develop sepsis. The main novel finding of this study is the significant association of the clinical diagnosis of sepsis with the continuous increase in the PSP level in the 3 days before this diagnosis. This association corroborates the initial description of this particularity of PSP [9], recently confirmed in a cohort of patients with severe burns, in which PSP levels increased constantly in the days preceding sepsis and this increase was greater in patients with septic shock [11, 23].
In addition, this study confirmed that the accuracy of clinical sepsis diagnosis based on PSP values was similar to that of diagnoses based on PCT and CRP values. Cut-off values for PSP the day of sepsis were however higher than usually reported in the literature, which is certainly due to the characteristics of the patients we included that differ from other studies with PSP in which it was evaluated at time of sepsis or suspected sepsis at admission. In an unselected population of critically ill patients, the use of PSP values to distinguish sepsis from non-infectious systemic inflammatory response syndrome (SIRS) at the time of ICU or high-dependency care admission was more accurate than the use of PCT values [10] was. Furthermore, a recent meta-analysis of data from 631 ICU patients, of whom 371 had infections or sepsis, confirmed that these conditions could be identified accurately based on PSP values [AUROCC = 0.81, slightly superior to those for PCT (0.78) and CRP (0.77)]. The combined use of PSP and CRP values increased this accuracy (AUROCC = 0.90) (Prazak J. et al., e-ISICEM, Sept. 15-18.2020), similarly to what was observed in this study where the combination of CRP and PSP increased the accuracy to 0.79, although the overlap of the 95% confidence intervals preclude to draw a robust statement regarding the significance of this increase.
This study has several strengths. First, it is the first international multicentric prospective study in which a host protein blood biomarker, the PSP, was measured daily in an ICU setting for the early detection of nosocomial sepsis using a point-of-care device offering on-demand test results in less than 10 min. Second, nosocomial sepsis was diagnosed by an independent adjudication committee composed of intensivists blinded to biomarker results. Third, this study determined the accuracy of single and combined biomarkers in the diagnosis of nosocomial sepsis in critically ill patients. Fourth, we explored the performance of the biomarkers, but we could not strictly interpret the value of bedside measurements of PSP compared to test performed in the clinical laboratory. However, several limitations of this study must be considered. Given its multicentric international design, the recruitment rate and sepsis incidence differed among study sites, in association with differences in preventive measures against nosocomial infection and sepsis development. In addition, although the expected number of patients who developed sepsis was reached, the low absolute number of sepsis events precluded subgroup analysis by site or country. Finally, this study was conducted with a subgroup of ICU patients who were expected to have prolonged ICU stays but were admitted free of infection, which precludes the extension of our observations to other settings, in particular for lower risk of developing nosocomial sepsis.
In the absence of an unambiguous definition of sepsis and highly accurate diagnostic tools the presence/absence of sepsis at any given days of an ICU stay was adjudicated by an expert committee. Discordant results using such approach have already been reported and vary considerably within sepsis diagnosis subgroups and clinical question [24], 25]. In our study not only the presence or absence of sepsis was adjudicated but also the day of event: this additional parameter contributed to the rate of discordance observed, reflecting the diagnostic challenges that clinicians are facing for the timely diagnosis of sepsis.
The inclusion of a risk-stratification of patients, for example based on clinical score such as the SOFA and the Acute Physiology And Chronic Health Evaluation (APACHE) scores, or of algorithms offering pretest probability in such diagnostic strategies, would allow to define economically viable and clinically meaningful strategies to closely monitor patients who will benefit the most of such approaches. In respect to that, the frequency of biomarkers measurement associated with the best clinical benefit remains to be established.
The clinical significance of an increase in the PSP level the days preceding the clinical diagnosis of sepsis may prompt changes in the management of patients at risk of nosocomial sepsis by triggering early diagnostic procedures and the timely establishment of appropriate treatment, for example to identify the source of the infection and the pathogen, and to assess the clinical utility of preemptive antibiotic therapy. Future interventional studies should also include a benefit-risk ratio of serial measurement of PSP, with a particular emphasis on the impact of false positive tests leading to unnecessary and potentially harmful diagnostic and therapeutic procedures. The results of this study may serve as a robust early basis for future validation studies of such an innovative approach, including studies conducted in wards outside of the ICU to determine how serial PSP measurement could enable early detection of sepsis before the onset of clinically overt symptoms in patients at risk.
Conclusions
The diagnostic accuracy of PSP, CRP and PCT for the diagnosis of sepsis at the time the EAC identified it were similar. Serial measurements of biomarker revealed that blood PSP levels increased incrementally 3 days before the clinical diagnosis of nosocomial sepsis in critically ill patients, potentially allowing the early detection of sepsis before the appearance of signs and symptoms. These results justify further evaluation of the potential of serial PSP measurement in the early diagnosis, management and clinical outcome of critically ill patients developing nosocomial sepsis.
Availability of data and materials
All data generated or analyzed during this study are included in this published article and its additional files.
Abbreviations
- APACHE:
-
Acute physiology and chronic health evaluation
- AUROCC:
-
Areas under receiver operating characteristic curves
- CI:
-
Confidence interval
- CRP:
-
C-reactive protein
- EAC:
-
External adjudication committee
- ICU:
-
Intensive care unit
- IQR:
-
Interquartile range
- IVD:
-
In vitro diagnostic
- PCT:
-
Procalcitonin
- PSP:
-
Pancreatic stone protein
- SIRS:
-
Systemic inflammatory response syndrome
- SOFA:
-
Sequential organ failure assessment
- ROC:
-
Receiver operating characteristic
References
Singer M, Deutschman CS, Seymour CW, et al. The third international consensus definitions for sepsis and septic shock (Sepsis-3). JAMA. 2016;315:801.
Seymour CW, Liu VX, Iwashyna TJ, et al. Assessment of clinical criteria for sepsis: for the third international consensus definitions for sepsis and septic shock (Sepsis-3). JAMA. 2016;315:762–74.
Levy MM, Evans LE, Rhodes A. The surviving sepsis campaign bundle: 2018 update. Intensive Care Med. 2018;44:925–8.
Rudd KE, Johnson SC, Agesa KM, et al. Global, regional, and national sepsis incidence and mortality, 1990–2017: analysis for the Global Burden of Disease Study. Lancet Lond Engl. 2020;395:200–11.
Husabø G, Nilsen RM, Flaatten H, et al. Early diagnosis of sepsis in emergency departments, time to treatment, and association with mortality: an observational study. PLoS ONE. 2020;15:e0227652.
Vincent J-L, Sakr Y, Sprung CL, et al. Sepsis in European intensive care units: results of the SOAP study. Crit Care Med. 2006;34:10.
Rhee C, Kadri SS, Danner RL, et al. Diagnosing sepsis is subjective and highly variable: a survey of intensivists using case vignettes. Crit Care Lond Engl. 2016;20:89.
Vijayan AL, Vanimaya N, Ravindran S, et al. Procalcitonin: a promising diagnostic marker for sepsis and antibiotic therapy. J Intensive Care. 2017;5:51.
Keel M, Seifert B, Bimmler D, et al. Pancreatic stone protein is highly increased during posttraumatic sepsis and activates neutrophil granulocytes. Crit Care Med. 2009;37:7.
Llewelyn MJ, Berger M, Gregory M, et al. Sepsis biomarkers in unselected patients on admission to intensive or high-dependency care. Crit Care. 2013;17:R60.
Klein HJ, Niggemann P, Buehler PK, et al. Pancreatic stone protein predicts sepsis in severely burned patients irrespective of trauma severity: a monocentric observational study. Ann Surg. 2020. https://doi.org/10.1097/SLA.0000000000003784.
Eggimann P, Que Y-A, Rebeaud F. Measurement of pancreatic stone protein in the identification and management of sepsis. Biomark Med. 2019;13:135–45.
Que Y-A, Delodder F, Guessous I, et al. Pancreatic stone protein as an early biomarker predicting mortality in a prospective cohort of patients with sepsis requiring ICU management. Crit Care. 2012;16:R114.
Que Y-A, Guessous I, Dupuis-Lozeron E, et al. Prognostication of mortality in critically 111 patients with severe infections. Chest. 2015;148:674–82.
García de Guadiana-Romualdo L, Albaladejo-Otón MD, Berger M, et al. Prognostic performance of pancreatic stone protein in critically ill patients with sepsis. Biomark Med. 2019;13:1469–80.
Putallaz L, Bogaard P van den, Laub P, et al.: Nanofluidics drives point-of-care technology for on the spot protein marker analysis with rapid actionable results. J Nanomed Nanotechnol. 2019; 10[cited 2021 Jan 7] Available from: https://www.longdom.org/open-access/nanofluidics-drives-pointofcare-technology-for-on-the-spot-protein-marker-analysis-with-rapid-actionable-results-44440.html
SepNet Critical Care Trials Group. Incidence of severe sepsis and septic shock in German intensive care units: the prospective, multicentre INSEP study. Intensive Care Med. 2016;42:1980–9.
Shankar-Hari M, Harrison DA, Rubenfeld GD, et al. Epidemiology of sepsis and septic shock in critical care units: comparison between sepsis-2 and sepsis-3 populations using a national critical care database. Br J Anaesth. 2017;119:626–36.
Artero A, Zaragoza R, Miguel J. Epidemiology of severe sepsis and septic shock. In: Fernandez R, editor. Severe sepsis and septic shock—understanding a serious killer. InTech; 2012.
Lambden S, Laterre PF, Levy MM, et al. The SOFA score-development, utility and challenges of accurate assessment in clinical trials. Crit Care Lond Engl. 2019;23:374.
Van Breukelen GJP. ANCOVA versus change from baseline: more power in randomized studies, more bias in nonrandomized studies [corrected]. J Clin Epidemiol. 2006;59:920–5.
Kahan BC. Accounting for centre-effects in multicentre trials with a binary outcome—when, why, and how? BMC Med Res Methodol. 2014;14:20.
Klein HJ, Rittirsch D, Buehler PK, et al. Response of routine inflammatory biomarkers and novel Pancreatic Stone Protein to inhalation injury and its interference with sepsis detection in severely burned patients. Burns. 2020; S0305417920303508
Klein Klouwenberg PMC, Ong DSY, Bos LDJ, et al. Interobserver agreement of Centers for Disease Control and Prevention criteria for classifying infections in critically ill patients. Crit Care Med. 2013;41:2373–8.
Lopansri BK, Miller Iii RR, Burke JP, et al. Physician agreement on the diagnosis of sepsis in the intensive care unit: estimation of concordance and analysis of underlying factors in a multicenter cohort. J Intensive Care. 2019;7:13.
Acknowledgements
We are grateful to all the physicians, nursing staff, and patients who participated in this study. In particular, we would like to thank: Aurélie Perret and Philippe Montillier at the Hôpitaux Universitaires de Genève, Geneva, Switzerland; Paul Bourzeix, Céline Prevost, Ludmila Baudrillart, Alexandra Gay, Anne-Aurore Duchambon, Fanny Mamert, Perrine Engel, Isabelle Herafa; Co-investigators : Philippe Vignon, Arnaud Desachy, Marine Goudelin, Anne Laure Fedou, Bruno Evrard, Pauline Segouffin at the Dupuytren Teaching Hospital, Limoges, France; Antonella Potalivo, Francesca Facondini and Andrea Pazzini at the Ospedale Infermi Rimini, Rimini, Italy; Francesco Anniciello and Alessandra Belli at the Azienda Ospedaliera di Perugia, Perugia, Italy; Cédric Darreau, Patrice Tirot, Mickael Landais, Nicolas Chudeau, Marjorie Saint-Martin, Jean-Christophe Callahan and Alain Robert at the Centre hospitalier Le Mans, Le Mans, France; Josef Prazak, Sabine Berger and the team of Research Nurses at the Bern University Hospital, Bern, Switzerland; Niall MacCallum, Deborah Smyth, Georgia Bercades and Jung Ryu at the University College Hospital, London, United Kingdom; James McKinlay, Kathryn McIntosh, Harry Brown, Gokce Daysal, Jessica Welch, Paula Carvelli, Humyraa Aziz, Laura Montague, Justin Kirk-Bayley and Lui Forni at the Royal Surrey County Hospital, Guildford, United Kingdom; Tiziana Casalicchio an Giulia Arena at the Ospedale San Giovanni Bosco, Torino, Italy; Maurizio Fusari and Filomena Di Antonio at Ospedale Santa Maria delle Croci, Ravenna, Italy. We also thank Dr. Antonio Perez for his help in the study conception and writing of the protocol and Laurie Girard for the coordination of this study.
Funding
The study was fully financed by Abionic SA (Epalinges, Switzerland; www.abionic.com), developer and manufacturer of the abioSCOPE®, a point-of-care CE-marked in vitro diagnostic device that uses proprietary nanofluidic technology to measure pancreatic stone protein. Fabien Rebeaud was involved in study design and writing of the manuscript.
Author information
Affiliations
Contributions
JP, PE and BF were involved in the conception and the design of the study. II performed the statistical analyses. JP, PE, BF, FR, II and TD analyzed the data and wrote the manuscript. All authors read and approved the final manuscript.
Corresponding author
Ethics declarations
Ethics approval and consent to participate
Approval was obtained from the ethics committees of all participating sites. The study has been registered at ClinicalTrials.gov (no. NCT03474809). Written informed consent to participation and any study-related assessment was obtained from all patients before their inclusion in the study. Written informed consent was obtained according to the specific requirements of each center’s ethics committee.
Consent for publication
N/A. The data presented are anonymized, so consent for publication is not required.
Competing interests
Irina Irincheeva is affiliated with CTU Bern, University of Bern, which has a staff policy of not accepting honoraria or consultancy fees. However, CTU Bern is involved in design, conduct, or analysis of clinical studies funded by not-for-profit and for-profit organizations. In particular, pharmaceutical and medical device companies provide direct funding to some of these studies. For an up-to-date list of CTU Bern’s conflicts of interest see http://www.ctu.unibe.ch/research/declaration_of_interest/index_eng.html. Pierre-François Dequin declares that his institution (Centre Hospitalier Régional Universitaire de Tours, Tours, France) received grants for the inclusion of patients in the study. Bruno François declares receiving consultancy fees from Abionic. The other authors declare that they have no conflicts of interest.
Additional information
Publisher's Note
Springer Nature remains neutral with regard to jurisdictional claims in published maps and institutional affiliations.
Supplementary Information
Additional file 1:
Table 1. (a) Study participating centers and (b) site recruitment. Table 2. Inclusion and exclusion criteria applied for patient recruitment. Table 3. Estimated coefficients of mixed-effects models for testing the consecutive increases in pancreatic stone protein, procalcitonin, and C-reactive protein levels serving as response variables. Table 4. Estimates for the mixed-effects models for which the results are reported in Table 3.
Additional file 2: Figure 1.
The abioSCOPE® device and its in vitro diagnostic CAPSULE pancreatic stone protein.
Additional file 3: Figure 2.
ROC curves for the diagnosis of sepsis at the time sepsis was clinically diagnosed by the EAC.
Rights and permissions
Open Access This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.
About this article
Cite this article
Pugin, J., Daix, T., Pagani, JL. et al. Serial measurement of pancreatic stone protein for the early detection of sepsis in intensive care unit patients: a prospective multicentric study. Crit Care 25, 151 (2021). https://doi.org/10.1186/s13054-021-03576-8
Received:
Accepted:
Published:
DOI: https://doi.org/10.1186/s13054-021-03576-8
Keywords
- Pancreatic stone protein
- Sepsis
- Diagnostic
- Procalcitonin
- C-reactive protein
- Biomarker