The dawn of physiological closed-loop ventilation—a review

A Letter to this article was published on 10 June 2020

Abstract

The level of automation in mechanical ventilation has been steadily increasing over the last few decades. There has recently been renewed interest in physiological closed-loop control of ventilation. The development of these systems has followed a similar path to that of manual clinical ventilation, starting with ensuring optimal gas exchange and shifting to the prevention of ventilator-induced lung injury. Systems currently aim to encompass both aspects, and early commercial systems are appearing. These developments remain unknown to many clinicians and, hence, limit their adoption into the clinical environment. This review shows the evolution of the physiological closed-loop control of mechanical ventilation.

Introduction

Closed-loop systems are designed to dynamically regulate a given variable around a desired set point. Examples thereof surround our everyday lives, from cruise-control maintaining the correct speed on the highway, to auto-pilot flying modern airplanes safely.

Modern medical systems are increasingly incorporating these technological advances. Medical applications of closed-loop control can be divided into systems that control a physical variable of the medical device and those that control a physiological variable of the patient [1]. Many medical devices already incorporate device-internal closed-loop control systems. An example is the internal regulation of the fraction of inspired oxygen to the value set by the clinician. If there is some backward interaction between the patient and the medical device, the system can be said to be patient-oriented. The control of airway pressure during pressure-controlled ventilation is such an example, because there is interaction between the device and patient. Here, the controller also focuses on the medical device.

Regulating a physiological variable of the patient is known as physiological closed-loop control (PCLC). With the extensive physiological monitoring in today’s clinical environment, PCLC is becoming ever more popular. In this case, the patient is the focus of the control loop and a physiological measurement is fed back to the controller. The PCLC has recently been taken up by regulatory bodies, with an international standard developed specifically for it, namely the IEC 60601-1-10 [2] and a public workshop held by the Food and Drug Administration in 2015 with subsequent publication by Parvinian and colleagues in 2018 [3].

Importantly, PCLC allows for the automation of certain therapeutic tasks currently performed by medical staff. Critical and emergency care especially is presumed to benefit from increased automation, as these high-stress environments have a shrinking workforce, as projected by Angus et al. [4]. This limited supply of medical staff, coupled with the labor-intensive practice of setting a ventilator correctly [5, 6] means that proper, personalized patient care will become even more difficult in the future. The already high costs of keeping patients on mechanical ventilation [7, 8] will also increase even further. If some ventilator settings are adjusted automatically, this would increase the time and capacity of the medical staff.

Despite regaining attention recently, PCLC of mechanical ventilation has been around for over half a century. Similar to the guidelines of mechanical ventilation which have evolved over time, from focusing on optimal gas exchange to reducing ventilator-induced lung injury [9], a similar trend can be seen in the research into PCLC for mechanical ventilation. This review aims to show the evolution of the PCLC of mechanical ventilation and its close relationship with the clinical goals of mechanical ventilation.

Scope and definitions

The criteria of search for this paper was as follows. Firstly, relevant literature was identified from database search combinations of the keywords: closed-loop, control, feedback and automation in combination with ventilation, mechanical ventilation, and artificial ventilation. Secondly, the search was narrowed down to include only systems which used patient-specific physiological variables for the feedback control. The physiological variables can be grouped loosely into oxygen, carbon dioxide, respiratory mechanics, and patient demand. Thirdly, the phase of weaning has so far benefited most from automation and was therefore added as an additional search keyword. Finally, only literature including studies with patients or medium/large animals was included. Extensive additional literature on theoretical approaches and simulative results exists, but these still need to be validated experimentally.

Furthermore, previous reviews on closed-loop mechanical ventilation exist and provided further relevant literature. Brunner presented an important early manuscript describing the history and principles of closed-loop ventilation [10]. Reviews of advanced closed-loop control in mechanical ventilation have been provided by Lellouche et al. [11], Wysocki et al. [12], and Branson [13].

In order to present a precise review, ventilation modes and breathing patterns are not discussed; these have been covered elsewhere, (e.g., Chatbrun et al. [14]). In addition, other methods of ventilation, such as liquid, noisy, and high-frequency ventilation, are beyond the scope of this paper.

Concept of physiological closed-loop ventilation

The primary goal of mechanical ventilation is to rectify and maintain adequate gas exchange. Ventilating patients protectively has become another important goal during mechanical ventilation.

The current clinical practice for choosing ventilator settings is very complex and based on expert knowledge. The modern clinician relies on classical physiologic variables for mechanical ventilation, such as peripheral capillary oxygen saturation (SpO2), end-tidal CO2 (etCO2), and blood gas analysis (PaO2, PaCO2, pH). In addition, clinicians consider the protective guidelines, hemodynamic parameters (heart rate, blood pressure, perfusion), and derived variables such as the pF ratio, the oxygen A-a gradient, shunt, transpulmonary pressure, and mechanical power, among others [9].

The interaction of clinician, ventilator, and patient forms a manual closed-loop system or “clinician-in-the-loop” system. This is shown in a block diagram in Fig. 1. Here the clinician acts as the controller and compares the available physiological and derived measurements to a defined control target before deciding on appropriate ventilator settings (actuation).

Fig. 1
figure 1

Classical clinician-in-the-loop system. The physiological measurement and ventilator settings shown are only exemplary. In the clinical environment, further derived measurement variables are also used. Clinician refers to the physicians, respiratory therapists, or nurses

This clinician-in-the-loop system is labor-intensive and time-consuming, as the presence of the clinician is always necessary. The clinician’s full attention is required to adjust ventilator settings if the patient state changes and to accommodate new therapeutic needs. If the clinician is not present, the system becomes an open-loop system, which is unable to respond if the oxygenation or ventilation become insufficient due to worsening patient conditions or external disturbances. This may lead to the patient being poorly and not protectively ventilated, with possible dire consequences.

Characteristics of closed-loop ventilation

An automated closed-loop system (also known as feedback control) can be implemented to keep a patient at a specified target and respond to disturbances without the clinician’s presence being necessary. Hereby, a controller takes over the task of adapting ventilator settings. The new control loop is depicted in Fig. 2 and shows the subtraction of a measured value from the target value and the error being fed into the controller. The controller then automatically decides on the correct ventilator settings to minimize the error.

Fig. 2
figure 2

Physiological closed-loop control for mechanical ventilation system

Importantly, the clinician’s focus changes to choosing personalized targets, regulating variables supplementary to the ventilation, such as hemodynamics and fluids, and monitoring the system. It should be noted that the current PCLC systems limit their measurement data to classical ventilation variables, as will be shown later. Derived evaluations, such as pF ratio, stress and strain, or heart-lung interactions, have not yet been used.

The goals of an automated closed-loop system can be divided into setpoint tracking and disturbance rejection. For most cases, setpoint control is only relevant when the controller is switched on, as it describes the dynamic response of the system until the target is reached. Setpoint changes are rare in the clinical environment. Exemplarily, the SpO2 target may be at 88–95% and seldom changes. The disturbance rejection, however, is the true hero during PCLC, as it is concerned with the response of the system to disturbances and bringing the patient back to the target range. Disturbances can take various forms, both internal (worsening illness, lung stiffness, increased CO2 production) and external (disconnection, room temperature). The two goals are shown for an illustrative example in Fig. 3, with a setpoint change at t1 and a disturbance (extreme increase in CO2) at t2.

Fig. 3
figure 3

Setpoint tracking and disturbance rejection shown for an illustrative example. A good controller ensures that the measured etCO2 closely follows the setpoint. At t1, a setpoint change (change in target) requires an increase in minute volume (bottom graph). At t2, a sudden increase in CO2 (disturbance) requires another increase in MV

The evolution of automated physiological control can be categorized by two drivers: the control target and the controller design. The control target is dictated on the one hand by the measurements and sensors available, some of which have been described above. On the other hand, changing clinical evidence and guidelines present new control targets. The second driver, controller design, remains an engineering problem and is beyond the scope of this paper. A brief overview is given here only to facilitate understanding. A natural evolution of the complexity of the controllers comes with the use of computers and their increasing computing power, and the detailed modeling of the respiratory system, which is used for controller synthesis and testing. Simple proportional, integral, and derivative (PID) controllers were used initially [15, 16]. This summation of PID action is the most common controller used in almost all industries nowadays. The tuning (setting up of this controller) is static, meaning the controller needs to be tuned to every patient. To overcome this, adaptive controllers [1719] can be used, which change their controller parameters to adapt to the patient or scenario. A mathematical model of the system is required for this, but recent advances in modeling respiratory systems have allowed their increased use. The optimal controller attempts to solve an optimization problem, such as minimizing a cost function, in calculating the actuation variable [20]. Model predictive control (MPC) is another example of an advanced controller which uses a dynamic model of the system [21, 22]. It optimizes the current ventilator settings for the current state, while anticipating future events. Finally, there has also been increasing use of fuzzy control [23, 24]. A fuzzy controller is a rule-based control law, in which the linguistic rules (e.g., if-then-else rules) are fuzzified (blurred) in order to make the control more elastic. These systems are popular in the clinical environment, as the language-based expert knowledge of clinicians is transferred into the design of the controller (which is not the case in PID control, for example). More robust controllers may be introduced in the future. These are designed with uncertainty in mind and can control the performance of the system even for the worst case.

History of physiological closed-loop control

The PCLC of ventilation covers a broad range of control targets. They can be grouped into controllers focusing on gas exchange, lung mechanics (protective ventilation), patient demand, and automation of clinical protocols.

Control based on gas exchange

The first application of PCLC in mechanical ventilation was presented by Saxton in 1953, with a publication appearing in 1957 [15], where his team applied feedback control to the iron lung to regulate the etCO2. In the same decade, Frumin developed an automated anesthesia system, which incorporated an etCO2 feedback control system [16, 25]. Both systems used the ventilation pressure as the actuating variable and were able to keep end-tidal CO2 at the set target.

In 1971, Mitamura and colleagues controlled the mixed expired CO2 using both tidal volume (VT) and breathing frequency (f) [20]. Their system was able to keep PCO2 at the target, even with extracorporeal CO2 loading. Other groups focused on setting either VT or f automatically, with the clinician setting the other variable manually [26, 27]. Digital control using computers for CO2 control started with Coles et al. in 1973 [26], and the number of publications about feedback control in ventilation increased. Coles et al. showed that the PCLC system maintained the etCO2 at the target better than manual control [26].

The availability of intravascular sensors for pH or PaCO2 measurement introduced new closed-loop control systems. Schulz et al. used such a sensor and feedback control to respond to setpoint changes in PaCO2 [28]. The response dynamics of their system was, however, limited by the slow response time of the sensor. A similar sensor problem caused the system of Coon et al. [29] to oscillate. The continuous intravascular sensors, however, failed to remain commercially available and their use has ceased. Without any clinical alternatives, the control of CO2 was again based on the measurement of etCO2. However, the increasing difference between PaCO2 and etCO2 in the pathological lung requires compensation [20, 28, 30]. Approaches based on first principles to link etCO2 and PaCO2 were presented by Ohlson et al. [30], but the authors could not compensate correctly when a large variation of cardiac output appeared.

An extensive list of CO2 feedback control (often referred to as ventilation control) systems over the past 50 years is shown in Table 1. As can be seen, all systems presented so far have been limited to proof-of-concept studies.

Table 1 Chronological development of closed-loop ventilation for CO2 and pH control in vivo

The focus was mostly on the control of CO2 until the early 1970s, but in 1975, Mitamura [36] developed a “dual control system” for both CO2 and O2. Here, the SpO2 was measured using an ear oximeter and an on-off controller was used for changing the inhaled oxygen content. The controller was able to rectify the hypoxia.

Controlling only the oxygenation was performed in preterm infants by Beddis et al. [37] in 1979; this was made possible by using an indwelling umbilical arterial oxygen electrode sensor. To evaluate their system, they compared the time spent at the oxygenation target using the controller to a clinician-in-the-loop system. This metric was also used by others [3844] and the automated system was as good or better than the manual procedure in all cases.

Yu et al. later used an oximeter on the tongue to control the oxygenation and implemented an adaptive controller in dogs [19]. This system was able to rectify hypoxia and compensate for disturbances, such as positive end-expiratory pressure (PEEP) changes and one lung ventilation. In 1991, East et al. [45] used both FiO2 and PEEP to control PaO2. Whether to change FiO2 or PEEP was based on previous clinical protocols and the system kept the patients at the oxygenation target for up to 6 h [45].

Further literature on oxygenation control is summarized in Table 2. The control of oxygenation using the fraction of inspired oxygen remains an active field of research, especially in neonates. A recent review of oxygenation control was published by Claure and Bancalari in 2013 [46]. The systems presented by Claure et al. [41], Urschitz et al. [42], and Gajdos et al. [44] have been developed further and are now commercially available as AVEA-CLiO2 (CareFusion, Yorba Linda, CA, USA), CLAC (Löwenstein Medical GmbH & Co. KG, Bad Ems, Germany), and SPOC (Fritz Stephan GmbH, Gackenbach, Germany), respectively.

Table 2 Chronological development of closed-loop ventilation for O2 control in vivo

Importantly, the control of oxygenation has mostly been limited to using only the FiO2 so far. This closely reflects the clinical difficulty in correctly choosing the PEEP—not least because oxygenation alone is not a reliable measurement of a good PEEP.

A large disadvantage of most of the closed-loop ventilation strategies presented is that they focus only on gas exchange and do not consider lung mechanics and quantification of harm. In fact, achieving proper CO2 control may require excessive VT or peak inspiration pressure, which can cause ventilator-induced lung injury (VILI). With clinical ventilation strategies becoming focused on being protective and preventing VILI, this requirement also needed to be incorporated into closed-loop control. Hence, control considering lung mechanics is presented next.

Control considering lung mechanics

Mitamura et al. considered minimizing ventilatory work as a further goal of their controller as early as 1971 [20]. The idea is closely related to that of Otis et al. [50] from 1950, which suggests that there exists an optimal combination of respiratory rate and tidal volume for minimal work of breathing. This approach was also used by Tehrani [51] and Laubscher [52] in 1991 and 1994, respectively. Laubscher et al. showed that their controller was able to adapt to personalized respiratory mechanics in a study on six patients. Laubscher and colleagues advanced their adaptive lung ventilation (ALV) controller (1994) to a newer version called adaptive support ventilation (ASV). Arnal et al. [53] tested the ASV controller on 243 patients with different respiratory lung conditions and showed the ability of the controller to choose VTf combinations related to actual personalized lung mechanics.

Rudowski et al. [54] addressed the concerns of VILI directly with the peak respiratory power index, as an index of lung trauma, in 1991. Their controller adjusts ventilator settings to reduce the respiratory power index, while ensuring adequate gas exchange. A study with six patients showed promising results.

Many modern systems do not directly control the ventilator using lung mechanics, but rather apply hard limits to ensure that tidal volume and pressure stay within certain, evidence-based, limits.

Control based on patient demand

It is also important to note here that the majority of literature presented so far considers only controlled (mandatory) ventilation, i.e., the patient is passive. However, in many cases, the patient is allowed or expected to breathe spontaneously, meaning patient demand becomes important. Early work was performed by Younes et al. with proportional assist ventilation (PAV) in 1992 [55]. This positive feedback control system amplified patient effort, according to the respiratory mechanics and level of assistance set by the operator. This ensures synchrony, while automatically adapting to patient load. For the initial version of PAV, the clinician required knowledge of the respiratory mechanics of the patient to set an appropriate controller gain, but the newer version, called PAV+, estimates the individual respiratory mechanics automatically [56, 57].

In 1996, Iotti et al. proposed P0.1 closed-loop control ventilation, whereby the drop in airway occlusion pressure during the first 0.1 s of inspiration is used to estimate patient work [58]. Two independent controllers, one for P0.1 and the other for alveolar volume, are fed into a merged control algorithm which changes the level of pressure support. The authors showed that inspiration activity of the patient can be stabilized at a desired level using P0.1, thus allowing for the unloading of the inspiratory muscles.

A direct coupling to the physiological neural output of the respiratory system would be helpful for optimal support during spontaneous breathing. An attempt to couple a respirator to phrenic nerve activity was performed in 1970 on animals [59], but this was not feasible in humans. Instead, Sinderby et al. [60] used the diaphragmatic electrical activity (EAdi) for neuro-ventilatory coupling to adjust the level of ventilatory support. This system requires the placement of an esophageal catheter. This system is commercially available as neurally adjusted ventilatory assist (NAVA) (Maquet Critical Care AB, Solna, Sweden). Improved patient-ventilator synchrony for this system was shown by [61], but the authors noted that the clinical impact thereof still needed to be determined.

Automation of clinical protocols

There have been various approaches to computerize clinical protocols which medical staff use to adapt mechanical ventilator settings. These computerized decision support systems do not make active changes to the ventilator but rather propose a change; the clinician has to be present to make the change and, as such, they represent classical clinician-in-the-loop systems. A literature review of these systems is given in [62].

Pomprapa et al. automated the ARDSNet protocol (autoARDSNet) and tested the system in a pilot study with seven pigs [63]. The system kept the animals at the oxygenation and pH targets for a duration of 4 h, even compensating for a disconnection between ventilator and patient.

Highly automated systems

Academia and industry have started to present highly and even fully automated systems which rely on a culmination of all the sub-categories of control targets presented above. This involves keeping an acceptable homeostasis of blood gases, ensuring patient-ventilator synchrony and preventing VILI. The topology for such a control loop is shown in Fig. 4.

Fig. 4
figure 4

Control topology for a fully automated physiological closed-loop ventilation. Measurement signals fed back to the controller are categorized according to the control target. The list of physiological measurements is not complete but shows only examples taken from the presented PCLC systems

The commercially available INTELLiVENT\(^{\circledR }\) - ASV\(^{\circledR }\) (Hamilton Medical AG, Bonaduz, Switzerland) is one example of a highly automated system. The control targets are the etCO2 and SpO2. In addition, the user has to provide some patient data, which is used for the estimation of lung mechanics. Oxygenation control is achieved by changing either PEEP or FiO2, according to the clinical guidelines from the ARDSNet protocol. Ventilation control (CO2 elimination) is controlled by a cascaded control loop, which includes the ASV controller described above.

Another highly automated system has been presented by Schwaiberger et al. in 2018 [64]. The etCO2 was controlled using fuzzy control; the principles of the open lung concept [65] were used for improved oxygenation and low VT ventilation ensuring protective ventilation was achieved [66]. A pilot study on eight pigs showed promising results but clinical trials are still pending. Furthermore, this system has only been applicable to passive patients.

Weaning

The final phase of mechanical ventilation has already seen some of the most advanced closed-loop systems becoming commercially available. The same control targets presented above also hold true for the weaning phase, but an additional goal during weaning is to reduce the reliance of the patient on the ventilator and to test whether the patient can be taken off mechanical aid completely.

A first approach to automated weaning was made by Hewlett et al. in 1977 [67] with mandatory minute ventilation.

However, the first automated physiological weaning approach was proposed by Chopin et al. in 1989 [68]. In their “carbon dioxide mandatory ventilation (CO2MV)” method, the authors used etCO2 and a rule-based controller to switch between spontaneous and controlled ventilation. In 1991, Strickland and Hasson [69] used pulse oximetry, respiratory rate, and minute ventilation to automatically set the synchronized intermittent mandatory ventilation rate and level of pressure support ventilation (PSV). Weaning was considered complete when the pressure support had reached zero; two or less mandatory breaths per minute were required and BGA showed sufficient evidence.

Dojat et al. [70] developed the GANESH system, which used the etCO2, respiratory rate, and VT as inputs for their rule-based controller to set the level of PSV. The goal of this strategy was to bring the patient into a zone of comfort, at which point permanent weaning was envisaged. This system was developed further, called NéoGanesh, and presented in 1997 [71] with a sophisticated knowledge and temporal reasoning controller. Finally, this system is available as SmartCare\(^{\circledR }\)/PS from Dräger (Drägerwerk AG, Lübeck, Germany).

The ALV controller by Laubscher et al. described previously was used for weaning by Linton and colleagues in 1994 [72]. Expanding on ALV, the ASV was developed. Given a patient capable of initiating spontaneous breathing, the ASV algorithm automatically decreases the inspiratory pressure. Once all breaths are spontaneous and stable gas exchange is ensured, weaning is considered complete. There was no direct feedback of gas exchange in the early versions. This has changed with the introduction of the INTELLiVENT\(^{\circledR }\) - ASV\(^{\circledR }\).

Evaluation of commercial physiological closed-loop ventilation

The papers presented so far have shown novel methods of applying physiological closed-loop control to mechanical ventilation and were able to test the systems on medium/large animals or a small number of patients. The commercial availability of some of these systems has allowed clinical studies to be performed. Brogi et al. did a systematic review and meta-analysis of closed-loop systems in the clinical environment. Five studies using automated FiO2 adjustments based on SpO2 measurements and three studies investigating ventilation control (etCO2) were included in the review. The time spent in the target zone was longer for automated systems in all studies [73]. The NAVA system was evaluated in a study by Demoule et al. and compared to the ventilation with PSV. Patient-ventilator asynchrony was reduced despite not increasing the probability of remaining in a partial ventilatory mode [74]. Rose et al. did a Cochrane systematic review and meta-analysis of clinical trials comparing automated and non-automated weaning [75]. The review included 21 trials, totaling 1676 patients, and included automated systems such as SmartCare\(^{\circledR }\)/PS, ALV, and INTELLiVENT\(^{\circledR }\) - ASV\(^{\circledR }\). They found that the weaning and ventilation duration was significantly reduced by automated systems [75]. Importantly, no strong effect on mortality was found. Burns et al. used the SmartCare system to compare closed-loop control of weaning to the manual protocol-based weaning [76]. They concluded that automated weaning showed promising results, but warranted further investigation [76]. A randomized controlled study by Lellouche et al. showed that the required number of interventions by clinical staff was reduced when using a fully automated ventilation system, thus, reducing the workload of the staff [77]. Arnal et al. showed that the INTELLiVENT\(^{\circledR }\) - ASV\(^{\circledR }\) reduced the number of manual ventilator setting changes in ICU patients [6]. They further concluded that this may increase the efficiency of the workforce [6].

These first analyses of clinical studies show promise that the application of PCLC to mechanical ventilation can reduce the workload of clinical staff and keep patients within personalized oxygenation and ventilation target zones safely. However, whether fully automated system will lead to improved mortality rates remains an open question.

Outlook

The evolution of the PCLC in mechanical ventilation continues to be driven by new controller designs, improved modeling, new sensor modalities, and better ventilation strategies.

Improved controller designs are needed that consider the cross-coupling effects of ventilator settings. Controlling CO2 by changing the MV will indirectly affect the oxygenation—which is not considered by any of the presented PCLC systems.

New measurements and imaging techniques are being applied in the “clinician-in-the-loop” setting, which could be used for automatic control in the future. The transpulmonary pressure, derived by subtracting the measured esophageal pressure from the airway pressure, is being researched as an approach to finding the appropriate PEEP setting [78]. Electrical impedance tomography (EIT), as a measurement technique, is gaining acceptance to titrate PEEP and also gain insight into the ventilated lungs. Recent reviews of electrical impedance tomography are presented by [7981].

Derived evaluations, such as the oxygen A-a gradient for oxygenation or stress and strain for protective ventilation, need to be incorporated into PCLC systems. The hemodynamic effects of ventilator changes (e.g., increasing PEEP) also need to be accounted for by the PCLC system. Furthermore, the medical history of the patient should be considered by the PCLC system.

With most data becoming digitized, it is plausible that this data will be available to the ventilator in the future. In such a cyber-medical world, all data will be transferred between all medical devices and this data will allow artificial intelligence to be applied to mechanical ventilation. Prasad and colleagues described a decision support system based on artificial intelligence [82].

Despite the many advantages that come with PCLC, the dangers thereof may not be forgotten. Sensor failure, unpredictable disturbances and software errors remain issues in any automated system and safety concepts need to be developed to ensure no harm is done to the patient. Kuck and Johnson [83] formulate the three laws for automation in anesthesia nicely: (1) do no harm, (2) be transparent, and (3) reduce the cognitive workload. A similar approach should be taken in the further development of closed-loop control of mechanical ventilation.

Conclusion

The application of PCLC to mechanical ventilation started well over half a century ago and all this research is beginning to bear fruit resulting in highly automated ventilators. As these first systems become commercially available, it is expected that more will follow, fed with new technology from industry and academia and, as such, the dawn of physiological closed-loop ventilation has certainly arrived.

Availability of data and materials

Not applicable.

Abbreviations

ALV:

Adaptive lung ventilation

ASV:

Adaptive support ventilation

BGA:

Blood gas analysis

CO2 :

Carbon dioxide

EAdi:

Diaphragmatic electrical activity

EIT:

Electrical impedance tomography

etCO2 :

End-tidal carbon dioxide

f:

Breathing frequency

FiCO2 :

Fraction inspired carbon dioxide

FiO2 :

Fraction inspired oxygen

I:

E: Inspiration to expiration ratio

MMV:

Mandatory minute ventilation

MV:

Minute ventilation

NAVA:

Neurally adjusted ventilatory assist

P insp :

Inspiratory pressure

PaCO2 :

Arterial carbon dioxide tension

PaO2 :

Arterial oxygen tension

PAV:

Proportional assist ventilation

PCLC:

Physiological closed-loop ventilation

PEEP:

Positive end-expiratory pressure

PID:

Proportional, integral, and differential

PIP:

Peak inspiratory pressure

PSV:

Pressure support ventilation

O2 :

Oxygen

SaO2 :

Arterial oxygen saturation

SIMV:

Synchronized intermittent mandatory ventilation

SpO2 :

Oxygen saturation (measured by oximetry)

tcO2 :

Transcutaneous oxygen tension

V T :

Tidal volume

VILI:

Ventilator-induced lung injury

References

  1. Walter M, Leonhardt S. Control applications in artificial ventilation. In: 2007 Mediterranean Conference on Control & Automation. Athens: IEEE: 2007. p. 1–6.

    Google Scholar 

  2. IEC 60601-1-10:2007 - Medical Electrical Equipment – Part 1-10: General requirements for basic safety and essential performance – collateral standard: requirements for the development of physiologic closed-loop controllers. https://www.iso.org/standard/44104.html.

  3. Parvinian B, Scully C, Wiyor H, Kumar A, Weininger S. Regulatory considerations for physiological closed-loop controlled medical devices used for automated critical care: food and drug administration workshop discussion topics. Anesth Analg. 2018; 126:1916–25.

    Article  PubMed  PubMed Central  Google Scholar 

  4. Angus DC, Kelley MA, Schmitz RJ, White A, Popovich Jr J, for the Committee on Manpower for Pulmonary, et al. Current and projected workforce requirements for care of the critically ill and patients with pulmonary disease can we meet the requirements of an aging population?JAMA. 284; 2000:2762.

  5. Vawdrey DK, Gardner RM, Evans RS, Orme JF, Clemmer TP, Greenway L, et al. Assessing data quality in manual entry of ventilator settings. J Am Med Inform Assoc. 2007; 14:295–303.

    Article  PubMed  PubMed Central  Google Scholar 

  6. Arnal JM, Garnero A, Novotni D, Corno G, Donati SY, Demory D, et al. Closed loop ventilation mode in intensive care unit: a randomized controlled clinical trial comparing the numbers of manual ventilator setting changes. Minerva Anestesiol. 2018; 84:58–67.

    PubMed  Google Scholar 

  7. Cox CE, Carson SS, Govert JA, Chelluri L, Sanders GD. An economic evaluation of prolonged mechanical ventilation. Crit Care Med. 2007; 35:1918–27.

    Article  PubMed  PubMed Central  Google Scholar 

  8. Zilberberg MD, de Wit M, Pirone JR, Shorr AF. Growth in adult prolonged acute mechanical ventilation: implications for healthcare delivery. Crit Care Med. 2008; 36:1451–5.

    Article  PubMed  Google Scholar 

  9. Pham T, Brochard LJ, Slutsky AS. Mechanical ventilation: state of the art. Mayo Clin Proc. 2017; 92:1382–400.

    Article  PubMed  Google Scholar 

  10. Brunner J. Principles and history of closed-loop controlled ventilation. Respir Care Clin N Am. 2001; 7:341–62.

    Article  CAS  PubMed  Google Scholar 

  11. Lellouche F, Brochard L. Advanced closed loops during mechanical ventilation (PAV, NAVA, ASV, SmartCare). Best Pract Res Clin Anaesthesiol. 2009; 23:81–93.

    Article  PubMed  Google Scholar 

  12. Wysocki M, Jouvet P, Jaber S. Closed loop mechanical ventilation. J Clin Monit Comput. 2014; 28:49–56.

    Article  PubMed  Google Scholar 

  13. Branson RD. Automation of mechanical ventilation. Crit Care Clin. 2018; 34:383–94.

    Article  PubMed  Google Scholar 

  14. Chatburn RL, El-Khatib M, Mireles-Cabodevila E. A taxonomy for mechanical ventilation: 10 fundamental maxims. Respir Care. 2014; 59:1747–63.

    Article  PubMed  Google Scholar 

  15. Saxton GA, Myers G. A servomechanism for automatic regulation of pulmonary ventilation. J Appl Physiol. 1957; 11:326–8.

    Article  PubMed  Google Scholar 

  16. Frumin MJ. Clinical use of a physiological respirator producing N2O amnesia-analgesia. Anesthes. 1957; 18:290–9.

    Article  CAS  Google Scholar 

  17. Takahara K, Wakamatsu H. Control of artificial respiration by adaptive pole-placement method. Syst Comput Jpn. 1994; 25:72–82.

    Article  Google Scholar 

  18. Sano A, Kikucki M. Adaptive control of arterial oxygen pressure of newborn infants under incubator oxygen treatments. IEE Proc D Control Theory Appl. 1985; 132:205.

    Article  Google Scholar 

  19. Yu C, He WG, So JM, Roy R, Kaufman H, Newell JC. Improvement in arteral oxygen control using multiple-model adaptive control procedures. IEEE Trans Biomed Eng. 1987; BME-34:567–74.

    Article  Google Scholar 

  20. Mitamura Y, Mikami T, Sugawara H, Yoshimoto C. An optimally controlled respirator. IEEE Trans Biomed Eng. 1971; BME-18:330–8.

    Article  Google Scholar 

  21. Fernando T, Cade J, Packer J. Automatic control of arterial carbon dioxide tension in mechanically ventilated patients. IEEE Trans Inform Technol Biomed. 2002; 6:269–76.

    Article  Google Scholar 

  22. Martinoni EP, Pfister CA, Stadler KS, Schumacher PM, Leibundgut D, Bouillon T, et al. Model-based control of mechanical ventilation: design and clinical validation. Br J Anaesth. 2004; 92:800–7.

    Article  CAS  PubMed  Google Scholar 

  23. Schäublin J, Derighetti M, Feigenwinter P, Petersen-Felix S, Zbinden AM. Fuzzy logic control of mechanical ventilation during anaesthesia. Br J Anaesth. 1996; 77:636–41.

    Article  PubMed  Google Scholar 

  24. Leonhardt S, Böhm S. Control methods for artificial ventilation of ARDS patients. In: European Medical & Biological Engineering Conference. Vienna: 1999. p. 2. https://www.tib.eu/en/search/id/TIBKAT%3A306783185/.

  25. Frumin MJ, Bergman NA, Holaday DA. Carbon dioxide and oxygen blood levels with a carbon dioxide controlled artificial respirator. Anesthesiology. 1959; 20:313–20.

    Article  CAS  PubMed  Google Scholar 

  26. Coles JR, Brown WA, Lampard DG. Computer control of respiration and anaesthesia. Med Biol Eng. 1973; 11:262–7.

    Article  CAS  PubMed  Google Scholar 

  27. Smith DM, Mercer RR, Eldridge FL. Servo control of end-tidal CO2 in paralyzed animals. J Appl Physiol. 1978; 45:133–6.

    Article  CAS  PubMed  Google Scholar 

  28. Schulz V, Ulmer HV, Erdmann W, Kunke S, Schnabel KH. Ein Verfahren zur paCO2-geregelten automatischen Ventilation. Pneumologie. 1974; 150:319–25.

    CAS  Google Scholar 

  29. Coon RL, Zuperku EJ, Kampine JP. Systemic arterial blood pH Servocontrol of mechanical ventilation. Anesthesiology. 1978; 49:201–4.

    Article  CAS  PubMed  Google Scholar 

  30. Ohlson KB, Westenskow DR, Jordan WS. A microprocessor based feedback controller for mechanical ventilation. Ann Biomed Eng. 1982; 10:35–48.

    Article  CAS  PubMed  Google Scholar 

  31. Holloman GH, Milhorn HT, Coleman TG. A sampled-data regulator for maintaining a constant alveolar CO2. J Appl Physiol. 1968; 25:463–8.

    Article  Google Scholar 

  32. East TD, Westenskow DR, Pace NL, Nelson LD. A microcomputer-based differential lung ventilation system. IEEE Trans Biomed Eng. 1982; BME-29:736–740.

    Article  Google Scholar 

  33. Bhansali PV, Rowley BA. A microcomputer-controlled servo ventilator. J Clin Eng. 1984; 9:47.

    Article  Google Scholar 

  34. Chapman FW, Newell JC, Roy RJ. A feedback controller for ventilatory therapy. Ann Biomed Eng. 1985; 13:359–72.

    Article  CAS  PubMed  Google Scholar 

  35. Ritchie RG, Ernst EA, Pate BL, Pearson JD, Sheppard LC. Closed-loop control of an anesthesia delivery system: development and animal testing. IEEE Trans Biomed Eng. 1987; BME-34:437–43.

    Article  Google Scholar 

  36. Mitamura Y, Mikami T, Yamamoto K. A dual control system for assisting respiration. Med Biol Eng. 1975; 13:846–54.

    Article  CAS  PubMed  Google Scholar 

  37. Beddis IR, Collins P, Levy NM, Godfrey S, Silverman M. New technique for servo-control of arterial oxygen tension in preterm infants. Arch Dis Child. 1979; 54:278–80.

    Article  CAS  PubMed  PubMed Central  Google Scholar 

  38. Dugdale RE, Cameron RG, Tealman GT. Closed-loop control of the partial pressure of arterial oxygen in neonates. Clin Phys Physiol Meas. 1988; 9:291–305.

    Article  CAS  PubMed  Google Scholar 

  39. Bhutani VK, Taube JC, Antunes MJ, Delivoria-Papadopoulos M. Adaptive control of inspired oxygen delivery to the neonate. Pediatr Pulmonol. 1992; 14:110–7.

    Article  CAS  PubMed  Google Scholar 

  40. Waisel DB, Fackler JC, Brunner JX, Kohane I. PEFIOS: An expert closed-loop oxygenation algorithm. Medinfo. 1995; 8 Pt 2:1132–36.

    CAS  PubMed  Google Scholar 

  41. Claure N, Gerhardt T, Everett R, Musante G, Herrera C, Bancalari E. Closed-loop controlled inspired oxygen concentration for mechanically ventilated very low birth weight infants with frequent episodes of hypoxemia. Pediatrics. 2001; 107:1120–4.

    Article  CAS  PubMed  Google Scholar 

  42. Urschitz MS, Horn W, Seyfang A, Hallenberger A, Herberts T, Miksch S, et al. Automatic control of the inspired oxygen fraction in preterm infants: a randomized crossover trial. Am J Respir Crit Care Med. 2004; 170:1095–100.

    Article  PubMed  Google Scholar 

  43. Morozoff PE, Evans RW, Smyth JA. Automatic control of blood oxygen saturation in premature infants. In: Proceedings of IEEE International Conference on Control and Applications. Vancouver: IEEE: 1993. p. 415–9.

    Google Scholar 

  44. Gajdos M, Waitz M, Mendler MR, Braun W, Hummler H. Effects of a new device for automated closed loop control of inspired oxygen concentration on fluctuations of arterial and different regional organ tissue oxygen saturations in preterm infants. Arch Dis Child Fetal Neonatal Ed. 2019; 104:F360–5.

    PubMed  Google Scholar 

  45. East TD, Tolle CRBS, MCJames SBS, Farrell RMBS, Brunner JX. A non-linear closed-loop controller for oxygenation based on a clinically proven fifth dimensional quality surface. Anesthesiol J Am Soc Anesthesiologists. 1991; 75(3):A468.

    Google Scholar 

  46. Claure N, Bancalari E. Automated closed loop control of inspired oxygen concentration. Respir Care. 2013; 58:151–61.

    Article  PubMed  Google Scholar 

  47. Raemer DB, Xin-Bao J, Topulos GP. FIx controller: an instrument to automatically adjust inspired oxygen fraction using feedback control from a pulse oximeter. J Clin Monit Comput. 1997; 13:91–101.

    Article  CAS  Google Scholar 

  48. Johannigman JA, Muskat P, Barnes S, Davis K, Beck G, Branson RD. Autonomous control of oxygenation. J Trauma Acute Care Surg. 2008; 64(4):S295–S301.

    Article  Google Scholar 

  49. Morozoff E, Smyth JA, Saif M. Applying computer models to realize closed-loop neonatal oxygen therapy. Anesth Analg. 2017; 124:95–103.

    Article  PubMed  Google Scholar 

  50. Otis AB, Fenn WO, Rahn H. Mechanics of Breathing in Man. J Appl Physiol. 1950; 2:592–607.

    Article  CAS  PubMed  Google Scholar 

  51. Tehrani FT. Automatic control of an artificial respirator. In: Proceedings of the Annual International Conference of the IEEE Engineering in Medicine and Biology Society Volume 13: 1991, Orlando, FL, USA,1991. p. 1738–9. https://ieeexplore.ieee.org/abstract/document/684729.

  52. Laubscher TP, Heinrichs W, Weiler N, Hartmann G, Brunner JX. An adaptive lung ventilation controller. IEEE Trans Biomed Eng. 1994; 41:51–59.

    Article  CAS  PubMed  Google Scholar 

  53. Arnal JM, Wysocki M, Nafati C, Donati S, Granier I, Corno G, et al. Automatic selection of breathing pattern using adaptive support ventilation. Intensive Care Med. 2008; 34:75–81.

    Article  PubMed  Google Scholar 

  54. Rudowski R, Bokliden A, Carstensen A, Gill H, Ludwigs U, Matell G. Multivariable optimization of mechanical ventilation. A linear programming approach. Int J Clin Monit Comput. 1991; 8:107–15.

    Article  CAS  PubMed  Google Scholar 

  55. Younes M. Proportional assist ventilation, a new approach to ventilatory support: theory. Am Rev Respir Dis. 1992; 145:114–20.

    Article  CAS  PubMed  Google Scholar 

  56. Younes M, Kun J, Masiowski B, Webster K, Roberts D. A method for noninvasive determination of inspiratory resistance during proportional assist ventilation. Am J Respir Crit Care Med. 2001; 163:829–39.

    Article  CAS  PubMed  Google Scholar 

  57. Younes M, Webster K, Kun J, Roberts D, Masiowski B. A method for measuring passive elastance during proportional assist ventilation. Am J Respir Crit Care Med. 2001; 164:50–60.

    Article  CAS  PubMed  Google Scholar 

  58. Iotti GA, Brunner JX, Braschi A, Laubscher T, Olivei MC, Palo A, et al. Closed-loop control of airway occlusion pressure at 0.1 second (P0.1) applied to pressure-support ventilation: algorithm and application in intubated patients. Crit Care Med. 1996; 24:771–9.

    Article  CAS  PubMed  Google Scholar 

  59. Remmers JE, Gautier H. Servo respirator constructed from a positive-pressure ventilator. J Appl Physiol. 1976; 41:252–5.

    Article  CAS  PubMed  Google Scholar 

  60. Sinderby C, Navalesi P, Beck J, Skrobik Y, Comtois N, Friberg S, et al. Neural control of mechanical ventilation in respiratory failure. Nat Med. 1999; 5:1433–6.

    Article  CAS  PubMed  Google Scholar 

  61. Piquilloud L, Vignaux L, Bialais E, Roeseler J, Sottiaux T, Laterre PF, et al. Neurally adjusted ventilatory assist improves patient–ventilator interaction. Intensiv Care Med. 2011; 37:263–71.

    Article  Google Scholar 

  62. Tehrani FT, Roum JH. Intelligent decision support systems for mechanical ventilation. Artif Intell Med. 2008; 44:171–82.

    Article  PubMed  Google Scholar 

  63. Pomprapa A, Schwaiberger D, Pickerodt P, Tjarks O, Lachmann B, Leonhardt S. Automatic protective ventilation using the ARDSNet protocol with the additional monitoring of electrical impedance tomography. Crit Care. 2014; 18:R128.

    Article  PubMed  PubMed Central  Google Scholar 

  64. Schwaiberger D, Pickerodt PA, Pomprapa A, Tjarks O, Kork F, Boemke W, et al. Closed-loop mechanical ventilation for lung injury: a novel physiological-feedback mode following the principles of the open lung concept. J Clin Monit Comput. 2018; 32:493–502.

    Article  PubMed  Google Scholar 

  65. Lachmann B. Open up the lung and keep the lung open. Intensiv Care Med. 1992; 18:319–21.

    Article  CAS  Google Scholar 

  66. Pomprapa A, Muanghong D, Köny M, Leonhardt S, Pickerodt P, Tjarks O, et al. Artificial intelligence for closed-loop ventilation therapy with hemodynamic control using the open lung Concept. Int J Intell Comput Cybern. 2015; 8:50–68.

    Article  Google Scholar 

  67. Hewlett AM, Platt AS, Terry VG. Mandatory minute volume.: a new concept in weaning from mechanical ventilation. Anaesthesia. 1977; 32:163–9.

    Article  CAS  PubMed  Google Scholar 

  68. Chopin C, Chambrin MC, Mangalaboyi J, Lestavel P, Fourrier F. Carbon dioxide mandatory ventilation (CO2MV): A new method for weaning from mechanical ventilation: description and comparative clinical study with I.M.V. and T. tube method in COPD patient. Int J Clin Monit Comput. 1989; 6:11–19.

    Article  CAS  PubMed  Google Scholar 

  69. Strickland JH, Hasson JH. A computer-controlled ventilator weaning system. Chest. 1991; 100:1096–9.

    Article  PubMed  Google Scholar 

  70. Dojat M, Brochard L, Lemaire F, Harf A. A knowledge-based system for assisted ventilation of patients in intensive care units. Int J Clin Monit Comput. 1992; 9:239–50.

    Article  CAS  PubMed  Google Scholar 

  71. Dojat M, Pachet F, Guessoum Z, Touchard D, Harf A, Brochard L. NéoGanesh: A working system for the automated control of assisted ventilation in ICUs. Artif Intell Med. 1997; 11:97–117.

    Article  CAS  PubMed  Google Scholar 

  72. Linton DM, Potgieter PD, Davis S, Fourie ATJ, Brunner JX, Laubscher TP. Automatic weaning from mechanical ventilation using an adaptive lung ventilation controller. Chest. 1994; 106:1843–50.

    Article  CAS  PubMed  Google Scholar 

  73. Brogi E, Cyr S, Kazan R, Giunta F, Hemmerling TM. Clinical performance and safety of closed-loop systems: a systematic review and meta-analysis of randomized controlled trials. Anesth Analg. 2017; 124:446–55.

    Article  PubMed  Google Scholar 

  74. Demoule A, Clavel M, Rolland-Debord C, Perbet S, Terzi N, Kouatchet A, et al. Neurally adjusted ventilatory assist as an alternative to pressure support ventilation in adults: a French multicentre randomized trial. Intensiv Care Med. 2016; 42:1723–32.

    Article  CAS  Google Scholar 

  75. Rose L, Schultz MJ, Cardwell CR, Jouvet P, McAuley DF. Blackwood B. Automated versus non-automated weaning for reducing the duration of mechanical ventilation for critically ill adults and children: a Cochrane systematic review and meta-analysis. Crit Care. 2015; 19:48.

    Article  PubMed  PubMed Central  Google Scholar 

  76. Burns KE, Lellouche F, Nisenbaum R, Lessard MR, Friedrich JO. Automated weaning and SBT systems versus non-automated weaning strategies for weaning time in invasively ventilated critically ill adults. Cochrane Database Syst Rev. 2014.

  77. Lellouche F, Bouchard PA, Simard S, L’Her E, Wysocki M. Evaluation of fully automated ventilation: a randomized controlled study in post-cardiac surgery patients. Intensiv Care Med. 2013; 39:463–71.

    Article  Google Scholar 

  78. Chiumello D, Guérin C. Understanding the setting of PEEP from esophageal pressure in patients with ARDS. Intensiv Care Med. 2015; 41:1465–7.

    Article  Google Scholar 

  79. Leonhardt S, Lachmann B. Electrical Impedance Tomography: The holy grail of ventilation and perfusion monitoring?Intensiv Care Med. 2012; 38:1917–29.

    Article  Google Scholar 

  80. Gong B, Krueger-Ziolek S, Moeller K, Schullcke B, Zhao Z. Electrical impedance tomography: functional lung imaging on its way to clinical practice?Expert Rev Respir Med. 2015; 9:721–737.

    Article  CAS  PubMed  Google Scholar 

  81. Frerichs I, Amato MBP, van Kaam AH, Tingay DG, Zhao Z, Grychtol B, et al. Chest electrical impedance tomography examination, data analysis, terminology, clinical use and recommendations: consensus statement of the TRanslational EIT developmeNt stuDy Group. Thorax. 2017; 72:83–93.

    Article  PubMed  Google Scholar 

  82. Prasad N, Cheng LF, Chivers C, Draugelis M, Engelhardt BE. A reinforcement learning approach to weaning of mechanical ventilation in intensive care units. 2017. arXiv:170406300 [cs].

  83. Kuck K, Johnson KB. The three laws of autonomous and closed-loop systems in anesthesia. Anesth Analg. 2017; 124:377–80.

    Article  PubMed  Google Scholar 

Download references

Acknowledgements

Not applicable.

Funding

Not applicable.

Author information

Authors and Affiliations

Authors

Contributions

PvP drafted the manuscript and composed the tables and images. AP, BL, and SL participated in defining the overall structure and selecting the literature to be included. All authors critically revised and approved the final manuscript.

Corresponding author

Correspondence to Philip von Platen.

Ethics declarations

Ethics approval and consent to participate

Not applicable.

Consent for publication

Not applicable.

Competing interests

The authors have cooperated and continued to work on closed-loop ventilation projects (SOLVe, AutoARDSNet, Oxyvent) funded by the Federal Ministry of Education and Research (BMBF, Germany).

Additional information

Publisher’s Note

Springer Nature remains neutral with regard to jurisdictional claims in published maps and institutional affiliations.

Rights and permissions

Open Access This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article’s Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article’s Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.

Reprints and permissions

About this article

Check for updates. Verify currency and authenticity via CrossMark

Cite this article

Platen, P.v., Pomprapa, A., Lachmann, B. et al. The dawn of physiological closed-loop ventilation—a review. Crit Care 24, 121 (2020). https://doi.org/10.1186/s13054-020-2810-1

Download citation

  • Received:

  • Accepted:

  • Published:

  • DOI: https://doi.org/10.1186/s13054-020-2810-1

Keywords