The usability of ventilators: a comparative evaluation of use safety and user experience

A Letter to this article was published on 06 July 2017

Abstract

Background

The design complexity of critical care ventilators (CCVs) can lead to use errors and patient harm. In this study, we present the results of a comparison of four CCVs from market leaders, using a rigorous methodology for the evaluation of use safety and user experience of medical devices.

Methods

We carried out a comparative usability study of four CCVs: Hamilton G5, Puritan Bennett 980, Maquet SERVO-U, and DrƤger Evita V500. Forty-eight critical care respiratory therapists participated in this fully counterbalanced, repeated measures study. Participants completed seven clinical scenarios composed of 16 tasks on each ventilator.

Use safety was measured by percentage of tasks with use errors or close calls (UE/CCs). User experience was measured by system usability and workload metrics, using the Post-Study System Usability Questionnaire (PSSUQ) and the National Aeronautics and Space Administration Task Load Index (NASA-TLX).

Results

Nine of 18 post hoc contrasts between pairs of ventilators were significant after Bonferroni correction, with effect sizes between 0.4 and 1.09 (Cohenā€™s d). There were significantly fewer UE/CCs with SERVO-U when compared to G5 (pā€‰=ā€‰0.044) and V500 (pā€‰=ā€‰0.020). Participants reported higher system usability for G5 when compared to PB980 (pā€‰=ā€‰0.035) and higher system usability for SERVO-U when compared to G5 (pā€‰<ā€‰0.001), PB980 (pā€‰<ā€‰0.001), and V500 (pā€‰<ā€‰0.001). Participants reported lower workload for G5 when compared to PB980 (pā€‰<ā€‰0.001) and lower workload for SERVO-U when compared to PB980 (pā€‰<ā€‰0.001) and V500 (pā€‰<ā€‰0.001). G5 scored better on two of nine possible comparisons; SERVO-U scored better on seven of nine possible comparisons. Aspects influencing participantsā€™ performance and perception include the low sensitivity of G5ā€™s touchscreen and the positive effect from the quality of SERVO-Uā€™s user interface design.

Conclusions

This study provides empirical evidence of how four ventilators from market leaders compare and highlights the importance of medical technology design. Within the boundaries of this study, we can infer that SERVO-U demonstrated the highest levels of use safety and user experience, followed by G5. Based on qualitative data, differences in outcomes could be explained by interaction design, quality of hardware components used in manufacturing, and influence of consumer product technology on usersā€™ expectations.

Background

Ventilators are a fundamental technology in critical care, with their use expected to increase in demand in the next 10Ā years [1]. Existing estimates of the proportion of patients admitted to the intensive care unit (ICU) requiring ventilator support range from 19Ā to 75Ā % in various countries [2ā€“5]. The use of ventilators is not without risk to the patient, with potential harm arising from infections, pneumothorax, ventilator-associated lung injury, and oxygen toxicity [6ā€“9].

Other significant ventilator-related risks are the associated use errors with the device [10ā€“14]. Use errors could cause patient harm in their operation if devices are not properly designed to mitigate such risks [11, 15, 16]. The design of ventilators can negatively affect user performance through poor user interfaces, interaction modes, or difficulties during the physical setup of the equipment [17].

The evaluation of use safety and the user experience of medical devices can be conducted through usability testing [18, 19]. Usability testing of medical devices has become increasingly important in recent years, with the US Food and Drug Administration (FDA) requiring medical devices to satisfy minimum use safety requirements prior to regulatory approval [20]. However, the testing conducted by manufacturers for their FDA submissions is often confidential, qualitative in nature, and not intended to achieve statistical significance in validating the product design [18]. As such, there are no means to compare the outcomes of these studies with those of similar devices on the market or with findings from other studies in the literature [21].

To address this limitation, a comparative usability test can be used, where multiple devices are evaluated concurrently, following the same protocol [18, 22]. Existing comparative studies available in the literature, however, normally test two devices or variations of a device design, as with, for example, the testing of compact transport ventilators [23], laparoscopic devices [24], and inhalers [25]. Other studies that provide comparisons of a larger number of devices rely on simplified methodologies that lack scientific rigor [26, 27].

In order to compare the use safety and user experience of critical care ventilators on the market, it was necessary to design and run a comparative usability test with a large enough sample of representative users to determine if there were statistically significant differences among ventilators.

This studyā€™s intent is to provide empirical evidence of the difference in use safety and user experience of four market-leading critical care ventilators available in North America [28]: the Hamilton G5 (Hamilton Medical AG; Bonaduz, Switzerland), the Covidien Puritan Bennett 980 (Covidien LP; Mansfield, MA, USA), the Maquet SERVO-U (Maquet Critical Care AB; Solna, Sweden), and the DrƤger Evita Infinity V500 (DrƤger Medical GmbH; LĆ¼beck, Germany). The findings explore the design of the ventilators in two dimensions of interest: use safety and user experience (a combination of perceived system usability and workload). The methodology presented in this paper enables users and decision makers to better understand the differences between designs of mechanical ventilators on the market, thereby supporting an understanding of user needs and procurement processes alike [29].

Methods

Experimental design

In order to estimate the sample size and the feasibility of a full-scale, powered study [30ā€“32], it was necessary to obtain a priori knowledge regarding the performance of the selected ventilators. As this data was not available in the literature, a pilot study with 13 participants was performed. Sample size was calculated assuming a repeated measures design with possible contrasts, with 80Ā % power and significance at the 5Ā % level [33] for use safety, system usability, and workload. Pilot data indicated a minimum of 48 participants was needed to discriminate between how each of the ventilators performed using the three selected metrics. This sample size allowed a full counterbalancing of the order of ventilators and the use of a repeated measures design to account for learning effects, order effects, and fatigue [30ā€“32]. Data analyses for the pilot study and the full-scale study were performed by the principal investigator, who was blinded to which dataset corresponded to which ventilator, to ensure objectivity.

The study was conducted at the Clinical Skills and Patient Simulation Center (CSPSC) at the University of North Carolina School of MedicineĀ (see Additional file 1) and received ethics approval (HHF1520_3) by the Quorum Review Independent Review Board (Quorum Review, Seattle, WA, USA). Written consent was collected from all study participants for the publication of their de-identified data and accompanying images used in this manuscript.

Participants

A total of 13 respiratory therapists (RTs) participated in the pilot study and 48 RTs in the full-scale study. RTs were chosen as the target group for this study because they are the primary daily users of ventilators in North American hospitals. The RT in the United States is responsible for ā€œclinical decision-making and patient education, [the RT] develops and implements respiratory care plans, applies patient-driven protocols, utilizes evidence-based clinical practice guidelines, and participates in health promotion, disease prevention, and disease management.ā€ [34]. RTs are responsible for responding in emergency situations, initiating and managing ventilators, and providing airway management in high-risk areas of North American hospitals, such as ICUs and emergency departments [34, 35].

Participants for this study were recruited from three hospital networks in North Carolina (Duke Health, WakeMed Health & Hospitals, and UNC Health Care). The CSPSC helped with recruitment by forwarding the recruitment email to head RTs in the three hospital networks, and a total of 143 RTs responded to the recruitment email. A survey was used during recruitment for selecting a group of participants with experience balanced among the four families of ventilators, to identify RTs with experience in critical care and to avoid recruiting participants with consulting relationships with manufacturers (see Additional file 2).

Devices

Each participant performed the scenarios on all four ventilators: the Hamilton G5 (G5), the Covidien Puritan Bennett 980 (PB980), the Maquet SERVO-U (SERVO-U), and the DrƤger Evita Infinity V500 (V500). The order of ventilators was fully counterbalanced, to avoid learning and order effects, and a repeated measures study design was used. These ventilators were selected for the study as the most advanced models from the North American market leaders [28].

Tasks and scenarios

The internationally recognized standard, ISO 80601-2-12 ā€” Particular requirements for basic safety and essential performance of critical care ventilators, details the primary operating functions in a ventilator-neutral and independent manner [36]. The 16 representative tasks described in TableĀ 1 were determined based on the aforementioned standard.

Table 1 Description of scenarios and associated tasks, as well as alarms and alerts

A total of seven scenarios were developed to create relevant clinical context, incorporating both typical clinical scenarios and time-sensitive scenarios, such as response to a loss of oxygen supply. The scenarios were designed by the authors and later vetted through consultations with RTs, ensuring accurate reflection of clinical context and patient conditions. The scenarios and tasks were always presented in the same order to maintain the clinical context, with the exception of the alarm tasks, which were randomized. All tasks had a 10-minute time limit, with a 160-minute time limit per ventilator.

In order to address concerns over instructor-led manufacturer training [37], this study employed exploration-based training [38] to increase realism and alignment with the delivery of training in real conditions (which was confirmed during the pilot study). Participants were given a set of learning objectives and asked to familiarize themselves with each of the four ventilators until they felt they would be comfortable using them on a real patient. A test administrator was available to answer questions and demonstrate any functions. At a minimum, all participants were required to demonstrate the ability to independently ventilate the patient, adjust parameters, adjust alarm limits, and browse menus.

Variables

Use safety, system usability, and workload were selected as the three major measures of interest, since they correlate to aspects of a medical deviceā€™s design quality and are known to affect patient safety [20, 39ā€“42]. These variables were measured through a combination of observed use and validated subjective scales, which measured participantsā€™ perception of the four devices and provided a comprehensive view of both the RTsā€™ perception and actual performance when using the ventilators. A combination of observed use and self-reported perception was used to protect this study from bias that may relate to the ā€œpreference versus performance paradoxā€ [43, 44].

Use safety was measured as inversely proportional to the percentage of tasks (total of 16) in which participants had a use error or close call (UE/CC); hence, a lower percentage represents a safer device. UE/CCs were collected through well-established observation techniques [18ā€“20]. A use error is defined as an action (or failure to act) that directly compromises safety or effectiveness of a device or that results in an undesirable or unintended treatment. A close call is defined as an instance in which a user experiences a usability issue that would result in a use error but successfully recovers prior to compromising the task. Two human factors experts independently observed each participant completing the tasks and recorded whether a UE/CC occurred during a task. This variable was categorical (UE/CC or no UE/CC), so a clinical task was completed either with a UE/CC or without one. Following data collection, observers compared ratings on a task-by-task basis and agreed on any clarifications to be sought during the participant debriefing. In cases where the issue remained unclear, a third human factors professional would independently resolve the tie through video review.

System usability was evaluated through a combination of the UE/CC metric described above and the Post-Study System Usability Questionnaire (PSSUQ) [45]: a 16-question, self-reported subjective evaluation of perceived system usability. The PSSUQ has been used in healthcare to evaluate clinical monitoring [46], anesthesia [47], as well as telerehabilitation systems [48]. PSSUQ scores range from 1 to 7, with lower scores representing better perceived system usability.

Lastly, workload was evaluated using the National Aeronautics and Space Administration Task Load Index (NASA-TLX) [49, 50], a subjective workload assessment tool that relies on six subscales (Mental Demand, Physical Demand, Temporal Demand, Own Performance, Effort, and Frustration). NASA-TLX has been used extensively in healthcare [50] to evaluate medical devices such as ventilators [51], infusion pumps [39], and physiological monitoring displays [52]. The output from the NASA-TLX instrument is a score ranging from 0 to 100, where lower scores correspond to lower perceived workload.

Data collection

Each participant session lasted a maximum of 8Ā hours, which included a 1.5-hour exploratory training period followed by four 1.5-hour periods (with breaks between periods), where participants performed the tasks on each of the four ventilators. For each ventilator, participants went through the scenarios described in TableĀ 1 while being observed by human factors experts responsible for logging the occurrence of UE/CCs and for collecting qualitative data about the performance of the participants. At the end of each sequence of scenarios, participants were presented with PSSUQ and NASA-TLX tools to evaluate the ventilator they had just used, followed by a debriefing interview to allow them to voice their opinions. This same process was repeated for each of the three remaining ventilators.

Data analysis

Statistical analyses were performed using SPSS Version 22.0 (IBM Corp, Armonk, NY, USA). Overall differences in safety and user experience of the ventilators were explored through repeated measures analysis of variance (ANOVA). Post hoc contrasts comparing any two ventilators were performed using multiple pairwise t tests. [30, 31]. Bonferroni corrections were used due to the multiple comparisons performed, with other less conservative corrections having minor effects on the number of statistically significant comparisons. Bonferroni correction can be overly conservative in studies of this type, considering that the outcome variables are correlated, increasing the chance of false negatives [53]. For this reason, the results of contrasts with Bonferroni correction and the post hoc t tests without correction are both reported in the results section (see TableĀ 4 for uncorrected results).

Results

A summary indicating how each pair of ventilators compares is presented in TableĀ 2, where only the statistically significant pair comparisons are presented. The SERVO-U outperformed the other ventilators in seven out of nine possible pair comparisons, and the G5 outperformed the other ventilators in two out of nine possible comparisons. The PB980 and the V500 did not outperform the other ventilators.

Table 2 Comparative description of how any two ventilators comparea

Overall ventilator comparison

TableĀ 3 outlines the percentage of tasks with UE/CCs, the perceived workload of each ventilator on the NASA-TLX scale, and the usability of the different ventilators as measured by the PSSUQ scale. Box plots, presented as an Additional file 3, provide a visual representation of these data.

Table 3 Ventilator performance in the UE/CC, NASA-TLX, and PSSUQ metrics

Repeated measures ANOVA showed statistically significant differences on all three variables: UE/CC, F(2.5, 119.1)ā€‰=ā€‰6.101, pā€‰<ā€‰0.001, partial Ī·2ā€‰=ā€‰0.115; NASA-TLX, F(3, 141)ā€‰=ā€‰16.629, pā€‰<ā€‰0.001, partial Ī·2ā€‰=ā€‰0.261; and PSSUQ, F(3, 141)ā€‰=ā€‰17.821, pā€‰<ā€‰0.001, partial Ī·2ā€‰=ā€‰0.275. Residuals were normally distributed.

Ventilator pair comparison

Six post hoc comparisons with Bonferroni correction [30, 31, 53, 54] were performed for each metric, which allowed pairs of ventilators to be ranked in terms of use safety (UE/CC), system usability (PSSUQ and UE/CC), and workload (NASA-TLX). The contrasts look at the differences in the means (MD) for each metric and determine, after corrections, whether these differences are statistically significant (TableĀ 4). After applying Bonferroni corrections, nine out of the 18 possible comparisons were statistically significant.

Table 4 Mean differences (MDā€‰ =ā€‰Vent1 ā€“ Vent2) between the ventilators with the results of the post hoc contrasts with Bonferroni correction (dfā€‰=ā€‰47), the post hoc t tests without corrections, and the effect sizes (Cohenā€™s d)

Participants experienced fewer UE/CCs with the SERVO-U (9.1Ā %) than with the G5 (12.8Ā %), MDā€‰=ā€‰āˆ’3.646, pā€‰=ā€‰0.044, dā€‰=ā€‰0.40. Participants also experienced fewer UE/CCs with the SERVO-U (9.1Ā %) than with the V500 (16.9Ā %), MDā€‰=ā€‰āˆ’7.813, pā€‰=ā€‰0.002, dā€‰=ā€‰0.55.

On the PSSUQ metric (ranging from 1 to 7), participants reported better usability for the G5 (2.7) than for the PB980 (3.5), MDā€‰=ā€‰āˆ’0.807, pā€‰=ā€‰0.035, dā€‰=ā€‰0.42. They also perceived better usability for the SERVO-U (1.7) compared to the G5 (2.7), PB980 (3.5), and V500 (3.1), MDā€‰=ā€‰āˆ’0.935, pā€‰<ā€‰0.001, dā€‰=ā€‰0.63; MDā€‰=ā€‰āˆ’1.742, pā€‰<ā€‰0.001, dā€‰=ā€‰1.07; MDā€‰=ā€‰āˆ’1.388, pā€‰<ā€‰0.001, dā€‰=ā€‰0.87, respectively.

Lastly, on the NASA-TLX metric (ranging from 0 to 100), participants reported lower workload for the G5 (28.3) compared to the PB980 (43.7), MDā€‰=ā€‰āˆ’15.449, pā€‰<ā€‰0.001, dā€‰=ā€‰0.64. They also reported lower workload for the SERVO-U (21.5) compared to the PB980 (43.7) and V500 (34.6), MDā€‰=ā€‰āˆ’22.214, pā€‰<ā€‰0.001, dā€‰=ā€‰1.09 and MDā€‰=ā€‰āˆ’13.144, pā€‰<ā€‰0.001, dā€‰=ā€‰0.62, respectively.

Effect sizes were within the 0.4 to 1.09 range, with most comparisons having medium (dā€‰>ā€‰0.5) and strong (dā€‰>ā€‰0.8) effect sizes (see TableĀ 4 for the complete results) [33].

Demographics

Data were collected from 48 participants for the full-scale study, out of which 34Ā % were male (nā€‰=ā€‰16) and 66Ā % were female (nā€‰=ā€‰32), with 68Ā % of the participants being between the ages of 25 and 45Ā years old (nā€‰=ā€‰33). As for experience, 63Ā % of the RTs who participated in the study (nā€‰=ā€‰30) had five or more years of experience as an RT.

A perfect balance of participantsā€™ level of experience with each of the ventilators was not possible due to uneven market share of the ventilators. However, using the data collected through the recruitment survey, multiple regression models were performed for all variables collected in the study, showing only minor effects on PSSUQ scores for the PB980, F(4,43)ā€‰=ā€‰4.796, pā€‰=ā€‰0.003, adj. R2ā€‰=ā€‰0.24, where only the experience with the PB980 (pā€‰=ā€‰0.044, Ī²ā€‰=ā€‰āˆ’0.268) and the G5 (pā€‰=ā€‰0.034, Ī²ā€‰=ā€‰āˆ’0.347) had an effect on the PSSUQ score for the PB980. All other variables collected in this study were not influenced by the experience with the ventilators.

Discussion

The intent of this study was to provide empirical evidence of the differences in use safety and user experience of four market-leading critical care ventilators available in North America. As the scenarios were the same for all four ventilators, the results presented in this paper suggest that the different user interfaces and interaction designs, as well as the quality of the hardware used, may have had an impact on user performance and perception. Additionally, the results reinforce the importance of user interfaces and user interaction in the design of medical technology [55ā€“57] as well as in the quality of the hardware used in manufacturing. For instance, the lack of sensitivity of the G5ā€™s touchscreen proved to be a barrier for task completion and a significant source of frustration, while the SERVO-Uā€™s user interface was praised by the participants. The design of a medical technology is a factor that can strongly influence user experience and user performance, as widely discussed in the medical device and critical care literature [55, 57ā€“59]. These results are also of critical importance for patient safety as they serve as an indicator of which medical technology is less likely to produce adverse events [55, 60, 61] arising from the operation of the devices.

The four ventilators were compared using repeated measures ANOVA, and we found statistically significant differences on all three variables (NASA-TLX, PSSUQ, UE/CC), showcasing medium (partial Ī·2ā€‰>ā€‰0.06) to large (partial Ī·2ā€‰<ā€‰0.13) effect sizes [33]. These results validate the sensitivity of our study design to discriminate the performance of the ventilators.

The participantsā€™ opinions were further supported by the results of the paired contrasts through repeated measures t test. The data from TableĀ 2 show that SERVO-U outperformed other ventilators in seven out of nine comparisons with other ventilators, showing medium to large effect sizes. These results indicate that participantsā€™ perceptions of the SERVO-Uā€™s superior user interface were reflected in the subjective and objective data collected in the study. SERVO-U showed safer performance (measured through UE/CC) when compared to the G5 and the V500, better perceived usability when compared to any of the three other ventilators, and lower perceived workload when compared to the PB980 and the V500. Next, the Hamilton G5 outperformed the PB980, both in self-reported usability and workload. The PB980 and the V500 did not outperform any ventilator in this study. Within the scope of this project, the SERVO-U, followed by the G5, demonstrated the highest levels of use safety and user experience, both factors that can directly impact patient safety [20, 40ā€“42].

Using only the quantitative results, it is not possible to ascertain which specific factors influenced user performance. Hence the importance of also collecting qualitative data in the form of observations to further enrich the analysis [18, 20]. The qualitative data also collected in this study indicate that the choices of interaction model of each ventilator (e.g., how to select information on the screen, adjusting settings, and confirming) seem to interfere with task completion and affect usersā€™ overall perception of the devices. A more detailed description of operational difficulties and safety implications of design should be explored in future publications, promoting an in-depth assessment of problems observed in this study.

The method used provided a comprehensive view of user experience and use safety of ventilators. NASA-TLX [49, 50], PSSUQ [45], and UE/CC [18ā€“20] have demonstrated their capacity for discriminating participantsā€™ performance on the ventilators, as well as for ranking the performance of medical devices available on the market. Even after applying Bonferroni corrections [53], our methodology was still able to discriminate the ventilators in 50Ā % of the possible comparisons (9/18 cases). In Europe, the tasks completed by RTs in this study are normally performed by nurses and doctors. Future studies could potentially compare the performance of RTs in North America with that of nurses and physicians in Europe.

Ultimately, the goal of this methodology is to support the design and/or selection of the safest medical devices on the market. The FDA, as well as researchers in patient safety, all posit the strong relationship that medical device usability has with use and patient safety [20, 40ā€“42], where devices with poor usability can potentially lead to harm to the patient. Hence, such a strong relationship should be reflected in our results. This effect was observed when comparing the SERVO-U with the V500 and G5 but not when comparing the SERVO-U to the PB980. This difference was a result of the conservative nature of Bonferroni corrections [53]. The uncorrected UE/CC comparison of the SERVO-U and PB980 is significant (see TableĀ 4), further supporting the relationship already discussed in the literature between usability and use safety.

In terms of further exploring the safety of medical technology, several studies in critical care that primarily focus on general characteristics and technical performance of medical devices would benefit from the rigorous methodology presented in this paper, to afford the evaluation of the human component on the use of technology, for example, studies of point-of-care technology [62] or emergency and transport ventilators [63], as well as those evaluating the effectiveness of electronic physician order entry in the ICU [59]. The effect of the human component has been extensively discussed in the critical care literature [58, 64, 65], describing how the design of humanā€“machine interfaces (or of medical device user interfaces) play an important role in the safety of critical care technology [56, 57].

Limitations of this study include the fidelity of simulated conditions and that only four ventilators were tested. Only RTs were included in the study, as opposed to nurses and physicians, who tend to be primary users outside North America. Additionally, the recruitment criteria and the structure of the demographic data limited our ability to run a regression analysis to evaluate the effect of different demographics variables on the variables being measured. Our study was not powered to run such regression analysis.

Lastly, this study was sponsored by the Maquet Getinge Group. Precautions and safeguards were taken to ensure the independence of the research. The study design, development of the methodology, selection of variables, data analysis, and manuscript preparation were made independently of the project sponsor. As we did not know how the ventilators would perform, a pilot study was used both for the calculation of sample size and to test the hypothesis that there would be measurable differences between ventilators. To further the independence of our research, all the statistical analyses were performed by the principal investigator, who was blinded to the identity of the ventilators.

Conclusions

This study provides empirical evidence on how the four ventilators from market leaders compare and highlights the importance of the design of medical technology. Within the boundaries of this study, we can infer that the SERVO-U ventilator demonstrated the highest levels of use safety and user experience, followed by the G5. Based on qualitative data collected during this study, differences in outcomes could be explained by interaction design, quality of the hardware components used in manufacturing, and influence of consumer product technology on usersā€™ expectations. Ultimately, the results presented in this paper provide evidence of the feasibility and potential of novel methodology comparative usability testing in identifying the safest and most usable medical technology on the market, supporting the selection of the safest medical technology and the design of the next generation of devices.

Abbreviations

ANOVA, analysis of variance; CSPSC, Clinical Skills and Patient Simulation Center; FDA, Food and Drug Administration; ICU, intensive care unit; NASA-TLX, National Aeronautics and Space Administration Task Load Index; PSSUQ, Post-Study System Usability Questionnaire; RT, respiratory therapist; UE/CC, use error/close call

References

  1. Needham DM, Bronskill SE, Calinawan JR, Sibbald WJ, Pronovost PJ, Laupacis A. Projected incidence of mechanical ventilation in Ontario to 2026: preparing for the aging baby boomers. Crit Care Med. 2005;33:574ā€“9.

    ArticleĀ  PubMedĀ  Google ScholarĀ 

  2. Edwards JR, Peterson KD, Andrus ML, Tolson JS, Goulding JS, Dudeck MA, et al. National Healthcare Safety Network (NHSN) Report, data summary for 2006, issued June 2007. Am J Infect Control. 2007;35:290ā€“301.

    ArticleĀ  PubMedĀ  Google ScholarĀ 

  3. Wunsch H, Wagner J, Herlim M, Chong D, Kramer A, Halpern SD. ICU occupancy and mechanical ventilator use in the United States. Crit Care Med. 2013;41:2712ā€“9.

    ArticleĀ  PubMedĀ  Google ScholarĀ 

  4. Esteban A, Anzueto A, AlĆ­a I, Gordo F, ApezteguĆ­a C, PĆ”lizas F, et al. How is mechanical ventilation employed in the intensive care unit? An international utilization review. Am J Respir Crit Care Med. 2000;161:1450ā€“8.

    ArticleĀ  CASĀ  PubMedĀ  Google ScholarĀ 

  5. Metnitz PGH, Metnitz B, Moreno RP, Bauer P, Sorbo LD, Hoermann C, et al. Epidemiology of mechanical ventilation: analysis of the SAPS 3 database. Intensive Care Med. 2009;35:816ā€“25.

    ArticleĀ  PubMedĀ  Google ScholarĀ 

  6. Marini JJ. Mechanical ventilation: past lessons and the near future. Crit Care. 2013;17:S1.

    ArticleĀ  PubMedĀ  PubMed CentralĀ  Google ScholarĀ 

  7. Michaud G, Cardinal P. Mechanisms of ventilator-induced lung injury: the clinicianā€™s perspective. Crit Care. 2003;7:209ā€“10.

    ArticleĀ  PubMedĀ  PubMed CentralĀ  Google ScholarĀ 

  8. Gilstrap D, Macintyre N. Patient ventilator interactions: implications for clinical management. Am J Respir Crit Care Med. 2013;188:1058ā€“68.

    ArticleĀ  PubMedĀ  Google ScholarĀ 

  9. Chastre J, Fagon J-Y. Ventilator-associated pneumonia. Am J Respir Crit Care Med. 2002;165:867ā€“903.

    ArticleĀ  PubMedĀ  Google ScholarĀ 

  10. Chatwin M, Heather S, Hanak A, Polkey MI, Simonds AK. Analysis of home support and ventilator malfunction in 1,211 ventilator-dependent patients. Eur Respir J. 2010;35:310ā€“6.

    ArticleĀ  CASĀ  PubMedĀ  Google ScholarĀ 

  11. Gravenstein JS. How does human error affect safety in anesthesia? Surg Oncol Clin N Am. 2000;9:81ā€“95.

    CASĀ  PubMedĀ  Google ScholarĀ 

  12. Auriant I, Reignier J, Pibarot ML, Bachat S, Tenaillon A, Raphael JC. Critical incidents related to invasive mechanical ventilation in the ICU: preliminary descriptive study. Intensive Care Med. 2002;28:452ā€“8.

    ArticleĀ  CASĀ  PubMedĀ  Google ScholarĀ 

  13. Craig J, Wilson ME. A Survey of anaesthetic misadventures. Anaesthesia. 1981;36:933ā€“6.

    ArticleĀ  CASĀ  PubMedĀ  Google ScholarĀ 

  14. ECRI Institute. Top 10 health technology hazards for 2016. 2015. https://www.ecri.org/Resources/Whitepapers_and_reports/2016_Top_10_Hazards_Executive_Brief.pdf. Accessed 14 July 2016.

    Google ScholarĀ 

  15. Gajic O, Frutos-Vivar F, Esteban A, Hubmayr RD, Anzueto A. Ventilator settings as a risk factor for acute respiratory distress syndrome in mechanically ventilated patients. Intensive Care Med. 2005;31:922ā€“6.

    ArticleĀ  PubMedĀ  Google ScholarĀ 

  16. Donchin Y, Seagull FJ. The hostile environment of the intensive care unit. Curr Opin Crit Care. 2002;8:316ā€“20.

    ArticleĀ  PubMedĀ  Google ScholarĀ 

  17. Cooper JB, Newbower RS, Long CD, McPeek B. Preventable anesthesia mishaps: a study of human factors. Qual Saf Health Care. 2002;11:277ā€“82.

    ArticleĀ  PubMedĀ  PubMed CentralĀ  Google ScholarĀ 

  18. Wiklund ME, Kendler J, Strochlic AY. Usability testing of medical devices. Boca Raton: CRC Press; 2011.

    Google ScholarĀ 

  19. Nielsen J. Usability engineering. Burlington: Morgan Kaufmann; 1993.

    Google ScholarĀ 

  20. FDA. Applying human factors and usability engineering to optimize medical device design. Food and Drug Administration. 2016. http://www.fda.gov/downloads/MedicalDevices/ā€¦/UCM259760.pdf. Accessed 14 July 2016.

  21. Ginsburg G. Human factors engineering: a tool for medical device evaluation in hospital procurement decision-making. J Biomed Inform. 2005;38:213ā€“9.

    ArticleĀ  PubMedĀ  Google ScholarĀ 

  22. Rubin J, Chisnell D. Handbook of usability testing: how to plan, design and conduct effective tests. Hoboken: Wiley; 2008.

    Google ScholarĀ 

  23. Templier F, Miroux P, Dolveck F, Descatha A, Goddet N-S, Jeleff C, et al. Evaluation of the ventilator-user interface of 2 new advanced compact transport ventilators. Respir Care. 2007;52:1701ā€“9.

    PubMedĀ  Google ScholarĀ 

  24. McCrory B, Lowndes BR, LaGrange CA, Miller EE, Hallbeck MS. Comparative usability testing of conventional and single incision laparoscopic surgery devices. Hum Factors. 2012;55:619ā€“31.

    ArticleĀ  Google ScholarĀ 

  25. Diggory P, Fernandez C, Humphrey A, Jones V, Murphy M. Comparison of elderly peopleā€™s technique in using two dry powder inhalers to deliver zanamivir: randomised controlled trial. Br Med J. 2001;322:577ā€“9.

    ArticleĀ  CASĀ  Google ScholarĀ 

  26. Gonzalez-Bermejo J, Laplanche V, Husseini FE, Duguet A, Derenne J-P, Similowski T. Evaluation of the user-friendliness of 11 home mechanical ventilators. Eur Respir J. 2006;27:1236ā€“43.

    ArticleĀ  CASĀ  PubMedĀ  Google ScholarĀ 

  27. Vignaux L, Tassaux D, Jolliet P. Evaluation of the user-friendliness of seven new generation intensive care ventilators. Intensive Care Med. 2009;35:1687ā€“91.

    ArticleĀ  PubMedĀ  Google ScholarĀ 

  28. Sichtermann L. Market analysis: ventilation. RT Mag. 2013. http://www.rtmagazine.com/2013/10/market-analysis-ventilation/. Accessed 14 July 2016.

  29. Cassano-PichĆ© A, Cafazzo JA, Chagpar A, Easty AC. How human factors methods can help in the procurement process. Biomed Instrum Technol. 2010;44:49ā€“56.

    ArticleĀ  Google ScholarĀ 

  30. Kutner MH, Nachtsheim CJ, Neter J. Applied linear statistical models. New York: McGraw-Hill/Irwin; 2005.

    Google ScholarĀ 

  31. Gravetter FJ, Wallnau LB. Statistics for the behavioral sciences. 7th ed. Belmont: Wadsworth Publishing; 2006.

    Google ScholarĀ 

  32. van Teijlingen ER, Hundley V. The importance of pilot studies. Soc Res Update. 2001;35:1ā€“4.

    Google ScholarĀ 

  33. Cohen J. Statistical power analysis for the behavioral sciences. London: Routledge; 1988.

    Google ScholarĀ 

  34. Commission on Accreditation of Allied Health Education Programs. Respiratory Therapist (Advanced). 2016. http://www.caahep.org/Content.aspx?ID=49. Accessed 14 July 2016.

  35. American Association of Respiratory Care. What is an RT? 2016. https://www.aarc.org/careers/what-is-an-rt/. Accessed 14 July 2016.

  36. ISO. ISO 80601-2-12:2011 Medical electrical equipment -- Part 2ā€“12: Particular requirements for basic safety and essential performance of critical care ventilators. 2011. http://www.iso.org/iso/catalogue_detail.htm?csnumber=51141. Accessed 14 July 2016.

  37. Richard JCM, Richard M, Kacmarek RM. ICU mechanical ventilators, technological advances vs. user friendliness: the right picture is worth a thousand numbers. Intensive Care Med. 2009;35:1662ā€“3.

    ArticleĀ  PubMedĀ  Google ScholarĀ 

  38. Kamouri AL, Kamouri J, Smith KH. Training by exploration: facilitating the transfer of procedural knowledge through analogical reasoning. Int J Man Mach Stud. 1986;24:171ā€“92.

    ArticleĀ  Google ScholarĀ 

  39. Carayon P, Wetterneck TB, Hundt AS, Ozkaynak M, Ram P, DeSilvey J, et al. Observing nurse interaction with infusion pump technologies. Adv Pat Saf. 2005;2:349ā€“64.

  40. Zhang J, Johnson TR, Patel VL, Paige DL, Kubose T. Using usability heuristics to evaluate patient safety of medical devices. J Biomed Inform. 2003;36:23ā€“30.

    ArticleĀ  PubMedĀ  Google ScholarĀ 

  41. Middleton B, Bloomrosen M, Dente MA, Hashmat B, Koppel R, Overhage JM, et al. Enhancing patient safety and quality of care by improving the usability of electronic health record systems: recommendations from AMIA. J Am Med Informatics Assoc. 2013;20:e2ā€“8.

    ArticleĀ  Google ScholarĀ 

  42. Karsh B-T. Beyond usability: designing effective technology implementation systems to promote patient safety. Qual Saf Health Care. 2004;13:388ā€“94.

    ArticleĀ  PubMedĀ  PubMed CentralĀ  Google ScholarĀ 

  43. Morita PP, Cafazzo JA. Challenges and paradoxes of human factors in health technology design. JMIR J Hum Factors. 2015;3:1ā€“7.

    Google ScholarĀ 

  44. Nielsen J, Levy J. Measuring usability: preference vs. performance. Commun ACM. 1994;37:66ā€“75.

    ArticleĀ  Google ScholarĀ 

  45. Lewis JR. Psychometric evaluation of the PSSUQ using data from five years of usability studies. Int J Hum Comput Interact. 2002;14:463ā€“88.

    ArticleĀ  Google ScholarĀ 

  46. Dosani M, Hunc K, Dumont GA, Dunsmuir D, Barralon P, Schwarz SKW, et al. A vibro-tactile display for clinical monitoring. Anesth Analg. 2012;115:588ā€“94.

    PubMedĀ  Google ScholarĀ 

  47. Gƶrges M, Winton P, Koval V, Lim J, Stinson J, Choi PT, et al. An evaluation of an expert system for detecting critical events during anesthesia in a human patient simulator. Anesth Analg. 2013;117:380ā€“91.

    ArticleĀ  PubMedĀ  Google ScholarĀ 

  48. Schutte JL, Gales S, Filipponi A, Saptono A, Parmanto B, McCue M. Evaluation of a telerehabilitation system for community-based rehabilitation system. Int J Telerehabilitation. 2012;4:15ā€“24.

    ArticleĀ  Google ScholarĀ 

  49. Hart SG, Staveland LE. Development of NASA-TLX (Task Load Index): results of empirical and theoretical research. Adv Psychol. 1988;1:139ā€“83.

    ArticleĀ  Google ScholarĀ 

  50. Hart SG. Nasa-Task Load Index (NASA-TLX); 20 years later. Proc Hum Factors Ergon Soc Annu Meet. 2006;50:904ā€“8. SAGE Publications.

  51. Wachter SB, Johnson K, Albert R, Syroid N, Drews F, Westenskow D. The evaluation of a pulmonary display to detect adverse respiratory events using high resolution human simulator. J Am Med Inform Assoc. 2006;13:635ā€“42.

    ArticleĀ  PubMedĀ  PubMed CentralĀ  Google ScholarĀ 

  52. Gƶrges M, Staggers N. Evaluations of physiological monitoring displays: a systematic review. J Clin Monit Comput. 2008;22:45ā€“66.

    ArticleĀ  PubMedĀ  Google ScholarĀ 

  53. Perneger TV. Whatā€™s wrong with Bonferroni adjustments. BMJ. 1998;316:1236ā€“8.

    ArticleĀ  CASĀ  PubMedĀ  PubMed CentralĀ  Google ScholarĀ 

  54. Sankoh AJ, Huque MF, Dubey SD. Some comments on frequently used multiple endpoint adjustment methods in clinical trials. Stat Med. 1997;16:2529ā€“42.

    ArticleĀ  CASĀ  PubMedĀ  Google ScholarĀ 

  55. Fairbanks RJ, Caplan S. Poor interface design and lack of usability testing facilitate medical error. Jt Comm J Qual Saf. 2004;30:579ā€“84.

    ArticleĀ  PubMedĀ  Google ScholarĀ 

  56. Hodge E, Blunt M, Young P. Capnography activation is improved by better ventilator interface ergonomics. Crit Care. 2011;15:P139.

    ArticleĀ  PubMed CentralĀ  Google ScholarĀ 

  57. Brindley PG. Patient safety and acute care medicine: lessons for the future, insights from the past. Crit Care. 2010;14:217.

    ArticleĀ  PubMedĀ  PubMed CentralĀ  Google ScholarĀ 

  58. Pickering BW, Litell JM, Herasevich V, Gajic O. Clinical review: the hospital of the future - building intelligent environments to facilitate safe and effective acute care delivery. Crit Care. 2012;16:220.

    ArticleĀ  PubMedĀ  PubMed CentralĀ  Google ScholarĀ 

  59. Colpaert K, Claus B, Somers A, Vandewoude K, Robays H, Decruyenaere J. Impact of computerized physician order entry on medication prescription errors in the intensive care unit: a controlled cross-sectional trial. Crit Care. 2006;10:R21.

    ArticleĀ  PubMedĀ  PubMed CentralĀ  Google ScholarĀ 

  60. Lin L, Vicente K, Doyle D. Patient safety, potential adverse drug events, and medical device design: a human factors engineering approach. J Biomed Inform. 2001;34:274ā€“84.

    ArticleĀ  CASĀ  PubMedĀ  Google ScholarĀ 

  61. Graham TAD, Kushniruk AW, Bullard MJ, Holroyd BR, Meurer DP, Rowe BH. How usability of a web-based clinical decision support system has the potential to contribute to adverse medical events. AMIA Annu Symp Proc. 2008;257ā€“61.

  62. Stadlbauer V, Wallner S, Sojakovic T, Smolle KH. Comparison of three different multi-analyte point-of-care devices during clinical routine on a medical ICU. Crit Care. 2011;433:e1ā€“11.

    Google ScholarĀ 

  63. Lā€™Her E, Roy A, Marjanovic N. Bench-test comparison of 26 emergency and transport ventilators. Crit Care. 2014;18:506.

    ArticleĀ  PubMedĀ  PubMed CentralĀ  Google ScholarĀ 

  64. Hugonnet S, Uckay I, Pittet D. Staffing level: a determinant of late-onset ventilator-associated pneumonia. Crit Care. 2007;11:1ā€“7.

    ArticleĀ  Google ScholarĀ 

  65. Sevdalis N, Brett SJ. Improving care by understanding the way we work: human factors and behavioural science in the context of intensive care. Crit Care. 2009;13:139.

    ArticleĀ  PubMedĀ  PubMed CentralĀ  Google ScholarĀ 

Download references

Acknowledgements

Our team would like to thank the team from the Clinical Skills and Patient Simulation Center at the University of North Carolina School of Medicine for all the support provided during this study, Meaghan Cuerden Knight for her input on statistical analyses, and Tara Fowler from Toronto General Hospital Respiratory Care for her support.

Funding

This study was sponsored by the Maquet Getinge Group.

Availability of data and materials

Not applicable.

Authorsā€™ contribution

PPM was the principal investigator and performed the statistical analyses, participant recruitment, study design, data analysis, data interpretation, and manuscript preparation. PBW contributed to the study design, participant recruitment, data acquisition, data analysis, data interpretation, and manuscript preparation. CJF contributed to the study design, participant recruitment, data acquisition, data analysis, data interpretation, and manuscript preparation. CAB contributed to the study design, participant recruitment, data acquisition, data analysis, data interpretation, and manuscript preparation. TAC contributed to participant recruitment, as well as data acquisition, data analysis, and data interpretation. MI contributed to participant recruitment, as well as data acquisition, data analysis, and data interpretation. AHP contributed to participant recruitment, as well as data acquisition, data analysis, and data interpretation. APS contributed to participant recruitment, as well as data acquisition, data analysis, and data interpretation. JAC contributed to the study design, data interpretation, and manuscript preparation. All authors reviewed, and approved the final manuscript.

Competing interests

All experimental design, data collection, data analysis, and manuscript preparation were performed without the influence of the sponsor. Final experimental design decisions and statistical analysis were performed by the studyā€™s principal investigator, who was blinded to which dataset corresponded to which ventilator. No other financial and non-financial conflicts of interest exist.

Consent for publication

Not applicable.

Ethics approval and consent to participate

This study received ethics approval (HHF1520_3) by the Quorum Review Independent Review Board (Quorum Review, Seattle, WA, USA). Written consent was collected from all study participants for the publication of their de-identified data and accompanying images used in this manuscript.

Author information

Authors and Affiliations

Authors

Corresponding author

Correspondence to Joseph A. Cafazzo.

Additional files

Additional file 1:

Setup of the testing facilities at the Clinical Skills and Patient Simulation Center at the University of North Carolina School of Medicine, where the simulator room and observation room can be seen. (PDF 1830Ā kb)

Additional file 2:

Recruitment survey used in this study. (PDF 641Ā kb)

Additional file 3:

Box plots showing the performance of the four ventilators in the use error/close call metric, the NASA-TLX scale, and the PSSUQ scale. Shading is used simply to differentiate datasets. Dots represent outliers. Lower scores on all three metrics correspond to better perception/performance. (PDF 72.4 kb)

Rights and permissions

Open Access This article is distributed under the terms of the Creative Commons Attribution 4.0 International License (http://creativecommons.org/licenses/by/4.0/), which permits unrestricted use, distribution, and reproduction in any medium, provided you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons license, and indicate if changes were made. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated.

Reprints and permissions

About this article

Check for updates. Verify currency and authenticity via CrossMark

Cite this article

Morita, P.P., Weinstein, P.B., Flewwelling, C.J. et al. The usability of ventilators: a comparative evaluation of use safety and user experience. Crit Care 20, 263 (2016). https://doi.org/10.1186/s13054-016-1431-1

Download citation

  • Received:

  • Accepted:

  • Published:

  • DOI: https://doi.org/10.1186/s13054-016-1431-1

Keywords