In response to the recent article in Critical Care by Morita et al. [1] entitled “The usability of ventilators: a comparative evaluation of use safety and user experience”, I agree that intensive care unit (ICU) safety and ventilator management are of paramount concern. The need for proper training and clinician education remains a constant priority for many ventilator manufacturers. To that end, the US Food and Drug Administration (FDA) now requires usability testing as part of the 510(k) clearance application [2], and I thank Morita et al. for addressing this important issue.

I would like to point out several discussion topics that I feel result in selection, but unintentional, bias to the study which may have resulted in the outcomes listed in the results of Morita et al. Selection bias occurs when researchers recruit an unrepresentative sample population. The sample population differs in some significant way from the population that generated the sample population [3]. Therefore, generalization of results should be made with caution. The test participants that were recruited were from three hospitals: Duke, Wake-Med, and UNC. These facilities currently utilize the Servo-I, PB840, and Hamilton G5 devices. Having the actual test device or immediate predecessor device of the study device provides test participants entering the testing process with advanced knowledge and experience versus that of equipment that is not used at their respective facilities.

Additionally, the study used “exploration-based training”. According to the FDA guidance on usability tests: “To the extent practicable, the content, format, and method of delivery of training given to test participants should be comparable to the training that actual users would receive”. Respiratory therapists (RTs) would typically receive training from the manufacturer or from manufacturer-certified trainers; therefore, the study should have offered such training before each test session [4]. Knowing that the study did not utilize manufacturer-trained clinical educators, the actual training used as the basis for the study must be questioned.

With these two points explained in the study design, I am of the opinion that the execution of the study with regard to recruited candidates and trainers introduces a level of bias that can easily skew the reported results [5]. I would suggest that a future study be performed with participants that have absolutely no prior experience of the study devices and that training be conducted by a certified manufacturer’s educator.

Thank you again to Morita et al. [1] for posting this study. ICU ventilator safety is of critical importance to everyone who plays a role in the delivery of care.

References

  1. Morita, et al. The usability of ventilators: a comparative evaluation of use safety and user Experience. Crit Care. 2016;20:263. doi:10.1186/s13054-016-1431-1.

    Article  PubMed  PubMed Central  Google Scholar 

  2. Usability.gov. Heuristic evaluations and expert reviews. http://www.usability.gov/how-to-and-tools/methods/heuristic-evaluation.html. Accessed 20 Oct 2014.

  3. Pannucci C, Wilkins E. Identifying and avoiding bias in research. Plastic Reconstructive Surg. 2010;126(2):619–22. doi:10.1097/PRS.0b013e3181de24bc.

    Article  CAS  Google Scholar 

  4. Branson, R. et al. Safe initiation and management of mechanical ventilation, American Association for Respiratory Care White Paper. https://www.aarc.org/wp-content/uploads/2016/05/White-Paper-SAFE-INITIATION-AND-MANAGEMENT-OF-MECHANICAL-VENTILATION.pdf.

  5. von Hippel PT. Mean, median and skew: correcting a textbook rule. J Statistics Educ. 2005;13(2):ww2.amstat.org/publications/jse/v13n2/vonhippel.html.

Download references

Acknowledgements

Not applicable.

Funding

None.

Availability of data and materials

Response letter only.

Authors’ contributions

Not applicable.

Authors’ information

EC is a Registered Respiratory Therapist with 30 years of experience and currently the Director of Marketing, Intensive Care, at Draeger, Inc.

Competing interests

EC is a full time employee of Draeger, Inc.

Consent for publication

Not applicable.

Ethics approval and consent to participate

Not applicable.

Declarations

None.

Publisher’s Note

Springer Nature remains neutral with regard to jurisdictional claims in published maps and institutional affiliations.

Author information

Authors and Affiliations

Authors

Corresponding author

Correspondence to Edwin Coombs.

Additional information

See related research by Morita et al., https://ccforum.biomedcentral.com/articles/10.1186/s13054-016-1431-1

Rights and permissions

Open Access This article is distributed under the terms of the Creative Commons Attribution 4.0 International License (http://creativecommons.org/licenses/by/4.0/), which permits unrestricted use, distribution, and reproduction in any medium, provided you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons license, and indicate if changes were made. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated.

Reprints and permissions

About this article

Check for updates. Verify currency and authenticity via CrossMark

Cite this article

Coombs, E. The usability of ventilators: a comparative evaluation of use safety and user experience—an editorial response. Crit Care 21, 164 (2017). https://doi.org/10.1186/s13054-017-1751-9

Download citation

  • Published:

  • DOI: https://doi.org/10.1186/s13054-017-1751-9