Skip to main content

Management and outcome of patients supported with Impella 5.0 for refractory cardiogenic shock

Abstract

Introduction

Cardiogenic shock refractory to standard therapy with inotropes and/or intra-aortic balloon pump is accompanied with an unacceptable high mortality. Percutaneous left ventricular assist devices may provide a survival benefit for these very sick patients. In this study, we describe our experience with the Impella 5.0 device used in the setting of refractory cardiogenic shock.

Methods

In this observational, retrospective, single-center study we included all the consecutive patients supported with Impella 5.0, between May 2008 and December 2013, for refractory cardiogenic shock. Patients’ baseline and procedural characteristics, hemodynamics and outcome to the first 48 h of support, to ICU discharge and day-28 visit were collected.

Results

A total of 40 patients were included in the study. Median age was 57 years and 87.5 % were male. Cardiogenic shock resulted from acute myocardial infarction in 17 patients (43 %), dilated cardiomyopathy in 12 (30 %) and postcardiotomy cardiac failure in 7 (18 %). In 15 patients Impella 5.0 was added to an ECMO to unload the left ventricle. The median SOFA score for the entire cohort prior to circulatory support was 12 [10–14] and the duration of Impella support was 7 [5–10] days. We observed a significant decrease of the inotrope score (10 [1–17] vs. 1 [0–9]; p = 0.04) and the lactate values (3.8 [1.7–5.9] mmol/L vs. 2.5 [1.5–3.4] mmol/L; p = 0.01) after 6 h of support with Impella 5.0. Furthermore, at Impella removal the patients’ left ventricular ejection fraction improved significantly (p < 0.001) when compared to baseline. Cardiac recovery, bridge to left ventricular assist device or heart transplantation was possible in 28 patients (70 %). Twenty-six patients (65 %) survived at day 28. A multivariate analysis showed a higher risk of mortality for patients with acute myocardial infarction (hazard ratio = 4.1 (1.2–14.2); p = 0.02).

Conclusions

Impella 5.0 allowed fast weaning of inotropes and might facilitate myocardial recovery. Despite high severity scores at admission, day-28 mortality rate was better than predicated.

Introduction

Cardiogenic shock (CS) remains a clinical challenge, with high mortality rates. Mechanical circulatory support (MCS) is increasingly used, whereas inotropes are associated with adverse events, inadequate circulatory supply or high myocardial oxygen consumption [1]. Data demonstrating the benefit of MCS in CS and helping guide the choice between the different available devices are needed [2]. A recent trial has demonstrated that intra-aortic balloon pump (IABP) did not improve outcomes of patients with CS complicating myocardial infarction (CSMI) [3, 4]. Peripheral venoarterial extracorporeal membrane oxygenation (PVA-ECMO) presents major advantages including quick and easy implantation, oxygen supply, biventricular assistance and good systemic blood flow; moreover, it is a cost-effective therapy. The PVA-ECMO can be used in various etiologies of CS [5] and seems to be the appropriate support for persistent cardiac arrest [6], biventricular failure and emergency MCS in remote institutions [7]. However, PVA-ECMO can be limited by insufficient left ventricular (LV) unloading resulting in thrombus formation or pulmonary edema, especially in case of profound LV failure [8, 9].

The Impella 5.0 device (Abiomed Europe GmbH, Aachen, Germany) provides a continuous blood flow up to 5 L/min and is designed to support patients in CS due to isolated LV failure for 10 days. Recent data have suggested that percutaneous left ventricular assist devices (LVADs) may provide a survival benefit for patients with refractory CS after myocardial infarction [10] or cardiotomy [11–13]. Experimental data have demonstrated that Impella 5.0 improved recovery after myocardial infarction [14]. Optimal LV unloading, which procures optimal myocardial oxygen supply/demand ratio, is better achieved with Impella, with or without associated PVA-ECMO [14, 15]. The use of Impella 2.5 devices (but not of Impella 5.0) for LV unloading in case of severe LV distension for patients on PVA-ECMO was previously described in case reports [16–19] and seems to be a promising therapeutic option.

The purpose of the current report was to describe our results with the Impella 5.0 device, used either as full LV support or complementary to PVA-ECMO in refractory CS, on cardiac recovery or successful switch to long-term LVAD or heart transplantation, ICU outcome, and day-28 mortality.

Materials and methods

Patients

We retrospectively reviewed the MCS database of our 14-bed intensive care unit (ICU) from May 2008 (first case of Impella 5.0 use) to December 2013 to register all patients implanted with Impella 5.0. During the study period, among 178 MCS cases recorded, the first-line support strategy for refractory CS was Impella 5.0 for 25 patients (14 %). The decision for support with Impella 5.0 first was taken by the physician on duty if the hemodynamic status of the patient with isolated LV failure was compatible with surgical implantation. The other 153 patients had had a PVA-ECMO first, and among them, LV unloading was required with Impella 2.5 (n = 14) or Impella 5.0 (n = 15; 10 % of PVA-ECMO patients). The cases with Impella 2.5 or PVA-ECMO alone are not described in this report.

The study was approved by the Montpellier University Hospital institutional review board, which, because of the retrospective nature of the study, waived the need for informed consent.

Definitions

Refractory CS was defined by evidence of tissue hypoxia (high blood lactate level, low venous oxygen saturation) concomitant with sustained hypotension and reduced cardiac index below 2.2 L/min/m2 despite adequate intravascular volume and optimal dose of inotropes and vasopressors. Isolated LV failure was defined by severely altered LV function (LV ejection fraction below 20 %) and no or mild right ventricular (RV) dysfunction at echocardiographic evaluation. Severe LV overload during PVA-ECMO support was considered in case of severe pulmonary edema, acute LV dilatation and/or visualization of spontaneous contrast in left heart cavities on echocardiography since PVA-ECMO support was set at the minimal level to cover metabolic needs as assessed by central venous oxygen saturation (ScvO2) and lactate. The beginning of MCS (t0MCS) matched with the time of first MCS implantation with either Impella 5.0 or PVA-ECMO. Sustained cardiac recovery was defined by a patient without inotropes or MCS or transplant at ICU discharge.

Mechanical circulatory support systems and implantation

All devices were implanted by an experienced cardiac surgeon. The Impella 5.0 device (Abiomed Europe GmbH, Aachen, Germany) is a 9 F catheter-mounted microaxial intracardiac pump (21 F), surgically inserted in the femoral artery (arteriotomy) or the right axillary artery (through a vascular graft) in the operating room or in the cardiac catheterization laboratory. It is positioned across the aortic valve into the left ventricle using fluoroscopy and/or transesophageal echocardiography and provides up to 5 L/min continuous blood flow by transvalvular active support with direct LV unloading. In Europe, this device is approved for short-term use, up to 10 days. PVA-ECMO was implanted in the operating room when transfer of the patient was possible, if not in the ICU or catheterization laboratory, and consisted of polyvinyl chloride tubing with a membrane oxygenator (PH.I.S.I.O and EOS; Sorin Group, Clamart, France), a centrifugal pump (Stockert; Sorin Group), and percutaneous or surgically inserted arterial and venous femoral cannulae (Fem-Flex and Fem-Track, Edwards Lifesciences, Guyancourt, France) with an additional 7 F cannula inserted distally into the femoral artery to prevent lower limb ischemia.

Patient management during Impella 5.0 support

Echocardiographic evaluation was systematically performed during initial settings, after each modification and before explantation of the device. Unfractionated heparin in hypertonic glucose solution (as recommended by the manufacturer) was administered through the Impella device to maintain an activated partial thromboplastin time between 1.5 and 2 times the normal value. For Impella-alone patients, the pump speed was adjusted to obtain sufficient blood flow for the oxygen demand with the lowest possible dose of inotropes. The RV function was assessed daily with echocardiography and enhanced if needed using vasopressors, inotropes, pacing, inhaled nitric oxide (iNO) to decrease RV afterload, and depletion. In case of persistent circulatory shock, especially because of RV failure, PVA-ECMO was added to the Impella device for global circulatory support. When Impella was secondarily combined with PVA-ECMO, the pump speed of Impella was adjusted in order both to unload LV and to prevent upper body hypoxia at the initial phase. After resolution of pulmonary edema, Impella flow was favored over PVA-ECMO flow and PVA-ECMO weaning was performed first, if possible.

The weaning process was left to the discretion of the physician guided by daily echocardiographic evaluation. Impella 5.0 weaning was performed in a stepwise fashion by decreasing the pump speed with echocardiography monitoring. Once the performance level of the device was reduced to the lowest level to avoid forward flow for 2 h, with stable cardiac index, mean arterial pressure (MAP) and ScvO2, the device was removed in the operating room. When MCS weaning was not possible, bridging to LVAD or heart transplantation was considered, provided there was no severe multiorgan failure.

Data collection

Demographic data and MCS characteristics were collected: age, sex, etiology of CS, Impella 5.0 implantation site, cardiac arrest before MCS, transfer on MCS from other centers, time from onset of CS to t0MCS, time from t0MCS to Impella 5.0 implantation for PVA-ECMO patients. The following clinical variables were collected at ICU admission and t0MCS: Simplified Acute Physiology Score II (SAPS II) [20], Sepsis-related Organ Failure Assessment (SOFA) score [21], MAP, invasive mechanical ventilation (MV) support, IABP support, inotropic support (dobutamine and/or epinephrine infusion) and blood lactate.

Before initiation, at ICU return (hour 0) and during the first 48 h of Impella 5.0 support, the following data were recorded: the vasoactive-inotropic score as defined as dose of dobutamine (μg/Kg/min) + [dose of epinephrine (μg/Kg/min) + dose of norepinephrine (μg/Kg/min)] x 100, and the inotrope score defined as dose of dobutamine (μg/Kg/min) + [dose of epinephrine (μg/Kg/min)] x 100 [22, 23], MV support, partial pressure of oxygen in arterial blood (PaO2) to inspired oxygen fraction (FiO2) ratio, MAP, blood lactate, N-terminal pro-brain natriuretic peptide (NT-proBNP) levels, Impella 5.0 and PVA-ECMO flows. Cardiothoracic ratios were calculated from digital chest radiographs (ratio between the maximum cardiac transverse diameter and the maximum thoracic diameter measured between the inner margins of the ribs).

Duration of Impella and PVA-ECMO support, occurrence of severe RV failure (needing additional PVA-ECMO) when Impella was used alone, PVA-ECMO weaning on Impella 5.0 support, MV weaning during the Impella course, incidence of transfusion and numbers of red blood cells (RBC) units transfused during MCS were also recorded.

Complications and outcome variables

Complications associated with Impella 5.0 were reported: major device malfunction (device failure), minor device malfunction (flow, position or pressure monitoring failure), device malposition (intra-aortic or intra-LV moving, successful bedside repositioning), incidence of blood transfusion during surgery, bleeding requiring reoperation, arterial ischemia at implantation site, stroke, hemolysis defined by increase of free bilirubin and lactate dehydrogenase (LDH) with or without anemia unexplained by bleeding (free hemoglobin dosage was not available at this time in our institution), ventricular arrhythmia, and device-related infection.

The main outcome variables included death before Impella removal, cardiac outcome in Impella survivors (sustained cardiac recovery; bridge to LVAD; bridge to heart transplantation), day-28 mortality, ICU mortality, and month-6 mortality. The other ICU outcome variables were the MV duration, the need for renal replacement therapy (RRT), and the length of ICU stay.

Statistical analysis

All data are presented as absolute values and percentages (%) for categorical variables or median and interquartile range (IQR) for continuous data. Paired Wilcoxon test was used to describe evolution of clinical and biological variables during first 48 h of Impella 5.0 support. Kaplan-Meier curves, with 28 days follow-up, were plotted to show the survival trend of patients according to the etiology of CS. Log-rank test was used to compare the survival difference between acute myocardial infarction and other etiologies. We performed a Cox proportional hazards regression model to identify variables that significantly influenced day-28 mortality, as measured by the hazard ratio with the 95 % confidence interval (CI). Variables related to mortality in the univariate analysis with p value < 0.2 were further analyzed in a stepwise multivariable Cox model. Statistical significance was defined as p value < 0.05. Analyses were performed using XLSTAT 2013 software (Addinsoft, New York, NY, USA).

Results

Study population

Baseline characteristics of the 40 patients (35 males, 87.5 %) who were supported by Impella 5.0 device for refractory CS are displayed in Table 1. Etiologies of CS were in order of frequencies: acute ST-elevation myocardial infarction (43 %), end-stage dilated cardiomyopathy (30 %), postcardiotomy (18 %) and others like myocarditis or contusion (10 %). Time from onset of CS to initiation of MCS (t0MCS) was 24 h (IQR 12–47). Twenty-five patients (62.5 % of the Impella 5.0 cohort) were treated by first intention with Impella 5.0 alone including three who had an additional PVA-ECMO. For one of them, PVA-ECMO was indicated for hypoxic cardiac arrest occurring during myocardial recovery and caused by acute respiratory distress syndrome in a patient with posttraumatic myocardial contusion. For the two other patients, MCS was upgraded for RV failure during Impella support (dilated cardiomyopathy and one myocarditis). Impella 5.0 was initiated in 15 patients (37.5 %) on PVA-ECMO support to unload the left ventricle and/or a switch strategy to a specific LV support. The median time between the beginning of extracorporeal support and Impella implantation was 20 h (IQR 7–45). Patients were younger but more severe at the time of MCS initiation in the subgroup of patients with Impella combined with PVA-ECMO versus patients supported by Impella first as assessed by SOFA score (respectively 14 [IQR 12.5–16] vs. 10 [IQR 8–12], p < 0.001) and blood lactate value (respectively 6 [IQR 5.1–12.6] vs. 3.4 [1.7–4.4], p = 0.008).

Table 1 Baseline characteristics of patients supported with the Impella 5.0 device

Mechanical circulatory support settings and clinical course

The flows of Impella and PVA-ECMO during the first 48 h were represented separately for Impella alone and Impella combined in Fig. 1. The ECMO flow was significantly reduced between hour 0 and hour 48 (3.9 [3.3–4.7] L/min versus 2.9 [2.4–3.6] L/min; p < 0.01) as the Impella flow tended to increase only when combined with PVA-ECMO (2.2 [1.8–3.1] L/min vs. 3.2 [2–3.9] L/min; p = 0.29). The clinical course, pump flows of MCS and hemodynamic parameters are summarized in Table 2. The use of inotropes was withdrawn early as assessed by the decrease of inotrope score from 10 [1–17] before implantation to 1 [0–9] at hour 6 (p = 0.04) while the blood lactate level decreased from 3.4 [1.7–5.1] mmol/L at hour 0 to 2.5 [1.5–3.4] mmol/L at hour 6 (p < 0.01). Significant improvement in the multiorgan dysfunction syndrome were observed during the first 2 days of Impella support as assessed by serial monitoring of the SOFA score (Table 2 and Fig. 2). Cardiac parameters of global overload like cardiothoracic ratio and NT-proBNP level were also decreased at hour 24 when compared to before Impella implantation (Table 2). Seven patients out of 25 on Impella 5.0 alone (28 %) were weaned from MV with a successful extubation during Impella support but none out of the 15 patients already on PVA-ECMO at the implantation time. Ten patients out of 15 (67 %) could be weaned from PVA-ECMO on Impella with a median time between Impella implantation and PVA-ECMO removal of 4 days (IQR 3–5).

Fig. 1
figure 1

Variations of mechanical circulatory support flow (median values) before and during the first 48 h of Impella support. Impella flow is represented by empty squares and ECMO flow by full circles. § p < 0.05 for comparison with H0 flow, paired Wilcoxon test. ECMO extracorporeal membrane oxygenation

Table 2 Clinical course and cardiac parameters before and during the first 48 h of support with Impella 5.0
Fig. 2
figure 2

Box plots of serial monitoring of the SOFA score, before mechanical circulatory support (t0MCS), during the first 48 h of Impella support and at the day of Impella removal. The line in the box indicates the median value of the data and the cross indicates the mean value. Paired Wilcoxon test for comparison with values at t0MCS and at day 0. SOFA Sepsis-related Organ Failure Assessment

Complications

Some complications were reported during Impella support (Table 3). In four patients, a major device malfunction resulted in stopping the circulatory support: one case at the ICU returned after PVA-ECMO removal with cardiac arrest despite no apparent anomaly on the Impella monitor (massive acute aortic regurgitation suspected, fatal issue), one case due to electric damage on the pump after transfer to the operating room for PVA-ECMO removal associated with hemorrhagic shock due to accidental ablation of the arterial cannula (fatal issue), and an unexplained technical problem for two of them (one after 5 h, the other one after 9 days of normal running) but nonfatal because of sufficient cardiac recovery. The other clinically relevant adverse events were: device-related infection (18 %), device displacement (20 %) and evidence or suspicion of hemolysis (10 %). The incidence of RBC transfusion was high: 30 patients (75 %) needed transfusion of at least one unit of RBC during the MCS period.

Table 3 Incidences of complications or adverse events during Impella 5.0 support

Clinical outcome

Left ventricular ejection fraction was significantly improved from 10 % (IQR 7–10) before implantation to 30 % (IQR 15–40) after Impella removal, p < 0.001 (Fig. 3). Cardiac outcome, main ICU management and mortality are reported in Table 4. Successful Impella weaning was achieved in 18 patients (45 % of total population, 64 % of survivors during Impella support) but two of them were implanted with an LVAD before ICU discharge because of remaining LV failure. In case of impossible weaning, direct transition to LVAD and heart transplantation was performed for seven and three patients respectively. None of the nine patients who underwent LVAD after Impella showed a RV failure after implantation.

Fig. 3
figure 3

Box plots of left ventricular ejection fraction measured before Impella 5.0 implantation and after Impella removal during ICU stay. The line in the box indicates the median value of the data and the cross indicates the mean value. Paired Wilcoxon test for comparison. ICU intensive care unit

Table 4 Cardiac, clinical and global outcomes

The mortality rate at day 28 was 35 % with a significant lower survival rate for CS complicating myocardial infarction when compared to other etiologies (47 % vs. 88 %; Khi2 test: p = 0.04), also shown on the Kaplan-Meier analysis (Fig. 4). No differences in outcome and in mortality were found between patients with Impella first and those with combination with PVA-ECMO.

Fig. 4
figure 4

Kaplan Meier analysis of day-28 survival for patients with cardiogenic shock complicating myocardial infarction versus other etiologies (dilated cardiomyopathy, postcardiotomy cardiogenic shock, miscellaneous)

Overall predictors of day-28 mortality

From the univariate analysis, SAPS II score at ICU admission, the rate of RBC transfusion by day on MCS and the acute myocardial infarction etiology were associated with day-28 mortality (Table 5). Stepwise multivariate model revealed that myocardial infarction etiology, vasoactive-inotropic score before Impella implantation and SAPS II score were independent risk factors for the day-28 mortality. Patients who had CS following acute myocardial infarction had a higher risk of mortality (hazard ratio = 4.1, 95 % CI (1.2–14.2); p = 0.02).

Table 5 Predictors of day-28 mortality in univariate and stepwise multivariate Cox regression analysis

Discussion

We report herein our experience with the Impella 5.0 device for patients in refractory CS of various etiologies, either as single LV support or as LV unloading on previous PVA-ECMO. To date, only a few studies have described the use of the Impella 5.0 device in CS, almost exclusively after cardiotomy [12, 13]. To our knowledge, it is the largest series of LV unloading with an Impella 5.0 device for patients on PVA-ECMO. Our single-center cohort of 40 Impella 5.0 implantations is original through its management of circulatory support going to a more selective strategy with an efficacy end point, which includes cardiac recovery or bridge to other therapy.

As a main result of this study, a low day-28 mortality rate of 35 % was observed in a population of refractory CS with significant and early improvement of LV ejection fraction in survivors after Impella removal when compared to before implantation. Second, Impella 5.0 allowed a sustained recovery of cardiac function in 40 % or a bridge to LVAD or urgent heart transplantation in 30 % of implanted patients. Finally, the Impella 5.0 device provided an efficient hemodynamic supply in refractory CS even if used alone and transition from PVA-ECMO to Impella 5.0 was possible in most patients with combined support.

In case of refractory CS, comparison between the different devices is difficult because of the lack of comparative trials. Published data [10, 24] suggest that survival was better when using Impella 5.0 compared to 2.5 devices for CS complicating myocardial infarction (CSMI). Recent studies [11–13] have reported a very good survival in case of postcardiotomy CS (PCCS) compared to historical series using Impella LD [25] or ECMO [26]. Less hemorrhagic complications and a better selection of patients more likely to benefit from this technique may explain these results. However, to date, the superiority of Impella 5.0 devices has not been proven and PVA-ECMO represents a more cost-effective device [27]. Whereas Impella 5.0 could be useful in selected patients presenting with CS due to isolated LV failure, the relative position of this device compared to other MCS systems needs to be clarified. PVA-ECMO is the unique option when RV function is impaired or in case of persistent cardiac arrest [6]. Because it allows full cardiac and respiratory supplies, PVA-ECMO also seems an appropriate choice for the most severe patients. However, an interesting potential advantage of Impella is the reduction of inotropes use. These treatments are associated with a worse prognosis of acute cardiac failure [28, 29]. In our series, the quick decrease of inotrope score was observed within 6 h after starting Impella and was withdrawn within 24 h for most patients. Another potential benefit of Impella is to unload the LV. Effective LV unloading without inotropes during Impella assistance was assessed by the decrease in NT-proBNP levels and cardiothoracic ratio within 24 h. We strongly believe that this LV unloading, which may prevent LV remodeling, and the restrictive use of inotropes are relevant advantages when compared to other techniques in this clinical setting. Improvement of myocardial recovery by Impella 5.0 device after acute myocardial infarction is already supported by experimental data [14, 30], especially versus PVA-ECMO [15]. Functional LV recovery was previously observed with Impella 2.5 in less severe LV failure [31] but clinical data with Impella 5.0 are missing. This is the first report that describes a major improvement of LV ejection fraction from very profound LV failure during Impella 5.0 support. Moreover, LV distension is a threatening complication of PVA-ECMO. It is the result of major LV failure and increased LV afterload because of the retrograde aortic reinjection of PVA-ECMO. The incidence and risk factors of this complication are not well described, in fact. Management of LV distention during PVA-ECMO is not standardized and various techniques have been described [8]. Because of complications with conventional techniques (atrial septotomy, ECMO centralization), interest in minimally invasive techniques such as IABP [32] and Impella devices is growing [16–19]. Using Impella 5.0 offers a larger range of flow than Impella 2.5, one reason to choose the largest device to unload the LV when a switch strategy was conceivable. Previous studies have already suggested a survival benefit of the use of Impella 5.0 compared to 2.5 devices for CSMI in small series [10, 24]. In our registry, LV distension occurred in the early phase of PVA-ECMO support as assessed by the period of 20 h between PVA-ECMO and Impella implantations. Moreover, successful transition from PVA-ECMO to Impella 5.0 support, achieved in 67 % cases including eight out of ten patients (80 %) who survived to short-term MCS, suggests that ECMO duration can be shortened while patients can still benefit from LV assistance with Impella. Median duration of PVA-ECMO was only 5 days, shorter than that described by Combes et al. (7 days for ICU survivors) in a comparable series of CS of various etiologies [5].

To date, few studies have described adverse events with Impella 5.0 device [11, 24]. However, as any effective MCS like IABP and ECMO, Impella 5.0 is an invasive technique requiring a surgical vascular approach of major arteries and is thus associated with several risks. Very few ischemic complications were observed, that is in agreement with the results of a meta-analysis on percutaneous LVAD [33]. A recurrent complication is bleeding and the increased need for RBC transfusion with a potential impact on mortality [34, 35]. As described by Loforte et al. [36], incidence of transfusion is high during PVA-ECMO, up to 100 %. In the present series, 75 % of patients were transfused. The high incidence of transfusion may be related to hemolysis and bleeding at the surgical site. Hemolysis occurred or was suspected in 10 % of our patients but was easily decreased when pump speed was reduced. Of note, a high incidence of device-related infection was observed, due to the bad clinical status of these patients, as usually observed in severe ICU patients. Not surprisingly, contamination of the device is then a possible complication, which was resolved after Impella removal in most cases. One major concern in this series is the occurrence of major device malfunctions leading to a fatal issue for two patients but some associated factors responsible for death were found: aortic regurgitation and hemorrhagic shock.

If the main goal during MCS for refractory CS remains the myocardium recovery, a bridge to LVAD can be a surrogate goal in younger patients when cardiac recovery is not possible. In this study, Impella 5.0 support allowed a bridge to LVAD for nine patients. Interestingly, none of these patients had presented a severe RV failure (defined by the need of RV mechanical support) after LVAD implantation. These results are surprisingly good when considering that about 20 % of patients requires RV support after LVAD implantation [37, 38]. In fact, if RV evaluation is crucial before LVAD implantation [39], based on hemodynamic and echocardiographic variables, it is a challenging problem in patients in CS because of the low cardiac output due to the LV failure or the unloaded RV by PVA-ECMO can conceal the ability of the RV to cope with the flow coming from a left ventricle assistance [40–42]. In this respect, Impella 5.0 support, as a first therapy or after PVA-ECMO, reproduces the real conditions of LVAD and allows an optimal evaluation of the RV function.

The day-28 mortality, the rate of ICU death and month-6 mortality were 35 %, 43 % and 50 % respectively, whereas severity scores at ICU admission predicated a mortality rate more than 60 %. A previous study in 2008 found an ICU mortality rate at 58 % for patients supported by ECMO for refractory CS [5]. Another retrospective study regarding less severe CS without initial indication for MCS except IABP reported a 30 % day-28 mortality rate [43]. Interestingly, our results confirm those previously published with Impella 5.0: two series in PCCS [12, 13], one in CSMI [10], one for CSMI and PCCS [11]. A higher risk of day-28 mortality was found in a multivariate analysis for patients with CSMI (Table 5). In a previous study with Impella 5.0, a lower survival rate was also reported for CS in the settings of myocardial infarction or dilated cardiomyopathy compared to PCCS [11]. One hypothesis to explain this difference should be the sudden and profound onset of CS following an extended myocardial infarction. The time from onset of CS to Impella 5.0 implantation is probably too long in these very poor circulatory conditions. A specific heart team with ICU physicians may provide an early recognition of patients rapidly progressing to refractory CS during percutaneous coronary interventions. High doses of vasopressors and inotropes before Impella implantation, as described by the vasoactive-inotropic score, were independently associated with higher mortality. This finding, supported by previous publications [11, 44], speaks in favor of MCS over escalating doses of medical treatment. To summarize about factors associated to mortality, SAPS II score at ICU admission and vasoactive-inotropic score before Impella are independently associated with mortality. We can speculate that Impella failure occurs more likely when shock is too much advanced. Although post-acute myocardial infarction cardiogenic shock is an independent factor associated with day-28 mortality, we can hypothesize that these patients have potential high benefits from support by Impella (high flow supply and LV unloading, with improved coronary blood flow and decreased myocardial workload). We strongly believe that an early decision for Impella implantation in this post-acute myocardial infarction population, before a high increase of vasopressors and inotropes treatment, may contribute to improved survival. This parameter should be taken into account in future comparative trials.

Our study has several limitations. First of all, as this is not a controlled trial with a comparison group versus Impella, all data presented can only help to generate hypotheses but cannot prove a survival benefit of Impella 5.0 support in these patients. Although the retrospective, single-center nature, and the sample size of this study are important limitations, to our knowledge, it is one of the largest series about Impella 5.0 utilization. The use of Impella 5.0 to unload the left ventricle associated with a transition strategy for patients on PVA-ECMO is a new description of ICU management during CS. Nevertheless, selection of patients who can benefit from this strategy among all patients on PVA-ECMO was not controlled and should be an important bias. Furthermore, although the indication of Impella 5.0 support in the first place was based on clinical and echocardiographic evaluation, some patients with isolated LV failure were probably supported with PVA-ECMO for many reasons like Impella monitor or operating room availabilities. Unfortunately, details on echocardiographic data like cardiac output and LV diameters are missing because echocardiographic examinations were not always securely recorded and only LV ejection fraction was systematically reported in medical charts. Another possible limitation is mixing various etiologies of cardiogenic shock, but the major characteristic of the hemodynamic failure originates from a primary cardiac failure; therefore, pump dysfunction is the common feature that requires similar treatment despite the various mechanisms.

Conclusions

This retrospective, observational study reports that Impella 5.0 support indicated for refractory cardiogenic shock was associated with a low day-28 mortality rate and led to myocardial recovery, transplantation or long-term support in 70 % of the implanted patients. Besides, some promising characteristics of Impella 5.0 support like fast weaning of inotropes, preservation of transpulmonary blood flow and optimal RV evaluation in the prospective of LVAD are suggested. Impella 5.0 seems a good mechanical assistance in the early phase of cardiogenic shock, when circulatory failure is not too critical. In acute ischemic conditions, it may protect against left ventricle remodeling, but further studies are needed to confirm long-term myocardial recovery and left ventricle function improvement.

Key messages

  • Day-28 mortality rate was 35 % after circulatory support with Impella 5.0 for refractory CS

  • Left ventricular ejection fraction improved from 10 to 30 % during Impella support

  • Sustained cardiac recovery or successful transition to LVAD or heart transplantation were obtained for 70 % of implanted patients

Abbreviations

CI:

confidence interval

CS:

cardiogenic shock

CSMI:

cardiogenic shock complicating myocardial infarction

FiO2 :

inspired oxygen fraction

IABP:

intra-aortic balloon pump

ICU:

intensive care unit

iNO:

inhaled nitric oxide

IQR:

interquartile range

LDH:

lactate dehydrogenase

LV:

left ventricular

LVAD:

left ventricular assist device

MAP:

mean arterial pressure

MCS:

mechanical circulatory support

MV:

mechanical ventilation

NT-proBNP:

N-terminal pro-brain natriuretic peptide

PaO2 :

partial pressure of oxygen in arterial blood

PCCS:

postcardiotomy cardiogenic shock

PVA-ECMO:

peripheral venoarterial extracorporeal membrane oxygenation

RBC:

red blood cells

RRT:

renal replacement therapy

RV:

right ventricular

SAPS II:

Simplified Acute Physiology Score II

ScvO2 :

central venous oxygen saturation

SOFA:

Sepsis-related Organ Failure Assessment

t0MCS:

time of first initiation of MCS

References

  1. McMurray JJ, Adamopoulos S, Anker SD, Auricchio A, Böhm M, Dickstein K, et al. ESC Guidelines for the diagnosis and treatment of acute and chronic heart failure 2012: The Task Force for the Diagnosis and Treatment of Acute and Chronic Heart Failure 2012 of the European Society of Cardiology. Developed in collaboration with the Heart Failure Association (HFA) of the ESC. Eur Heart J. 2012;33:1787–847.

    PubMed  Google Scholar 

  2. Werdan K, Gielen S, Ebelt H, Hochman JS. Mechanical circulatory support in cardiogenic shock. Eur Heart J. 2014;35:156–67.

    CAS  PubMed  Google Scholar 

  3. Thiele H, Zeymer U, Neumann FJ, Ferenc M, Olbrich HG, Hausleiter J, et al. Intraaortic balloon support for myocardial infarction with cardiogenic shock. N Engl J Med. 2012;367:1287–96.

    CAS  PubMed  Google Scholar 

  4. Thiele H, Zeymer U, Neumann FJ, Ferenc M, Olbrich HG, Hausleiter J, et al. Intra-aortic balloon counterpulsation in acute myocardial infarction complicated by cardiogenic shock (IABP-SHOCK II): final 12 month results of a randomized, open-label trial. Lancet. 2013;382:1638–45.

    PubMed  Google Scholar 

  5. Combes A, Leprince P, Luyt CE, Bonnet N, Trouillet JL, Léger P, et al. Outcomes and long-term quality-of-life of patients supported by extracorporeal membrane oxygenation for refractory cardiogenic shock. Crit Care Med. 2008;36:1404–11.

    PubMed  Google Scholar 

  6. Chen YS, Lin JW, Yu HY, Ko WJ, Jerng JS, Chang WT, et al. Cardiopulmonary resuscitation with assisted extracorporeal life-support versus conventional cardiopulmonary resuscitation in adults with in-hospital cardiac arrest: an observational study and propensity analysis. Lancet. 2008;372:554–61.

    PubMed  Google Scholar 

  7. Beurtheret S, Mordant P, Paoletti X, Marijon E, Celermajer DS, Léger P, et al. Emergency circulatory support in refractory cardiogenic shock patients in remote institutions: a pilot study (the cardiac-RESCUE program). Eur Heart J. 2013;34:112–20.

    PubMed  Google Scholar 

  8. Soleimani B, Pae WE. Management of left ventricular distension during peripheral extracorporeal membrane oxygenation for cardiogenic shock. Perfusion. 2012;27:326–31.

    CAS  PubMed  Google Scholar 

  9. Petroni T, Harrois A, Amour J, Lebreton G, Brechot N, Tanaka S, et al. Intra-aortic balloon pump effects on macrocirculation and microcirculation in cardiogenic shock patients supported by venoarterial extracorporeal membrane oxygenation. Crit Care Med. 2014;42:2075–82.

    CAS  PubMed  Google Scholar 

  10. Engstrom AE, Cocchieri R, Driessen AH, Sjauw KD, Vis MM, Baan J, et al. The Impella 2.5 and 5.0 devices for ST-elevation myocardial infarction patients presenting with severe and profound cardiogenic shock: the Academic Medical Center intensive care unit experience. Crit Care Med. 2011;39:2072–9.

    PubMed  Google Scholar 

  11. Lemaire A, Anderson MB, Lee LY, Scholz P, Prendergast T, Goodman A, et al. The Impella device for acute mechanical circulatory support in patients in cardiogenic shock. Ann Thorac Surg. 2014;97:133–8.

    PubMed  Google Scholar 

  12. Engstrom AE, Granfeldt H, Seybold-Epting W, Dahm M, Cocchieri R, Driessen AH, et al. Mechanical circulatory support with the Impella 5.0 device for postcardiotomy cardiogenic shock: a three-center experience. Minerva Cardioangiol. 2013;61:539–46.

    CAS  PubMed  Google Scholar 

  13. Griffith BP, Anderson MB, Samuels LE, Pae Jr WE, Naka Y, The FOH, et al. a multicenter prospective study of Impella 5.0/LD for postcardiotomy circulatory support. J Thorac Cardiovasc Surg. 2013;145:548–54.

    PubMed  Google Scholar 

  14. Wei X, Li T, Hagen B, Zhang P, Sanchez PG, Williams K, et al. Short-term mechanical unloading with left ventricular assist devices after acute myocardial infarction conserves calcium cycling and improves heart function. JACC Cardiovasc Interv. 2013;6:406–15.

    PubMed  Google Scholar 

  15. Kawashima D, Gojo S, Nishimura T, Itoda Y, Kitahori K, Motomura N, et al. Left ventricular mechanical support with Impella provides more ventricular unloading in heart failure than extracorporeal membrane oxygenation. ASAIO J. 2011;57:169–76.

    PubMed  Google Scholar 

  16. Vlasselaers D, Desmet M, Desmet L, Meyns B, Dens J. Ventricular unloading with a miniature axial flow pump in combination with extracorporeal membrane oxygenation. Intensive Care Med. 2006;32:329–33.

    PubMed  Google Scholar 

  17. Cheng A, Swartz MF, Massey HT. Impella to unload the left ventricle during peripheral extracorporeal membrane oxygenation. ASAIO J. 2013;59:533–6.

    CAS  PubMed  Google Scholar 

  18. Koeckert MS, Jorde UP, Naka Y, Moses JW, Takayama H. Impella LP 2.5 for left ventricular unloading during venoarterial extracorporeal membrane oxygenation support. J Card Surg. 2011;26:666–8.

    PubMed  Google Scholar 

  19. Beurtheret S, Mordant P, Pavie A, Leprince P. Impella and extracorporeal membrane oxygenation: a demanding combination. ASAIO J. 2012;58:291–3.

    PubMed  Google Scholar 

  20. Le Gall JR, Lemeshow S, Saulnier F. A new Simplified Acute Physiology Score (SAPS II) based on a European/North American multicenter study. JAMA. 1993;270:2957–63.

    PubMed  Google Scholar 

  21. Vincent JL, Moreno R, Takala S, De Mendonca A, Bruining H, Reinhart CK, et al. The SOFA (Sepsis-related Organ Failure Assessment) score to describe organ dysfunction/failure. On behalf of the Working Group on Sepsis-Related Problems of the European Society of Intensive Care Medicine. Intensive Care Med. 1996;22:707–10.

    CAS  PubMed  Google Scholar 

  22. Wernovsky G, Wypij D, Jonas RA, Mayer Jr JE, Hanley FL, Hickey PR, et al. Postoperative course and hemodynamic profile after the arterial switch operation in neonates and infants. A comparison of low-flow cardiopulmonary bypass and circulatory arrest. Circulation. 1995;92:2226–35.

    CAS  PubMed  Google Scholar 

  23. Davidson J, Tong S, Hancock H, Hauck A, da Cruz E, Kaufman J. Prospective validation of the vasoactive-inotropic score and correlation to short-term outcomes in neonates and infants after cardiothoracic surgery. Intensive Care Med. 2012;38:1184–90.

    PubMed  Google Scholar 

  24. Lauten A, Engstrom AE, Jung C, Empen K, Erne P, Cook S, et al. Percutaneous left-ventricular support with the Impella-2.5-assist device in acute cardiogenic shock: results of the Impella-EUROSHOCK-registry. Circ Heart Fail. 2013;6:23–30.

    PubMed  Google Scholar 

  25. Siegenthaler MP, Brehm K, Strecker T, Hanke T, Nötzold A, Olschewski M, et al. The Impella Recover microaxial left ventricular assist device reduces mortality for postcardiotomy failure: a three-center experience. J Thorac Cardiovasc Surg. 2004;127:812–22.

    CAS  PubMed  Google Scholar 

  26. Rastan AJ, Dege A, Mohr M, Doll N, Falk V, Walther T, et al. Early and late outcomes of 517 consecutive adult patients treated with extracorporeal membrane oxygenation for refractory postcardiotomy cardiogenic shock. J Thorac Cardiovasc Surg. 2010;139:302–11.

    PubMed  Google Scholar 

  27. Chamogeorgakis T, Rafael A, Shafii AE, Nagpal D, Pokersnik JA, Gonzalez-Stawinski GV. Which is better: a miniaturized percutaneous ventricular assist device or extracorporeal membrane oxygenation for patients with cardiogenic shock? ASAIO J. 2013;59:607–11.

    PubMed  Google Scholar 

  28. Tacon CL, McCaffrey J, Delaney A. Dobutamine for patients with severe heart failure: a systematic review and meta-analysis of randomized controlled trials. Intensive Care Med. 2012;38:359–67.

    CAS  PubMed  Google Scholar 

  29. Nielsen DV, Hansen MK, Johnsen SP, Hansen M, Hindsholm K, Jakobsen CJ, et al. Health outcomes with and without use of inotropic therapy in cardiac surgery: results of a propensity score-matched analysis. Anesthesiology. 2014;120:1098–108.

    PubMed  Google Scholar 

  30. Stamm C, Friehs I, Cowan DB, Cao-Danh H, Choi YH, Duebener LF, et al. Dopamine treatment of postischemic contractile dysfunction rapidly induces calcium-dependent pro-apoptotic signaling. Circulation. 2002;106:290–8.

    Google Scholar 

  31. Casassus F, Corre J, Leroux L, Chevalereau P, Fresselinat A, Seguy B, et al. The use of Impella 2.5 in severe refractory cardiogenic shock complicating an acute myocardial infarction. J Interv Cardiol. 2015;28:41–50.

    PubMed  Google Scholar 

  32. Ma P, Zhang Z, Song T, Yang Y, Meng G, Zhao J, et al. Combining ECMO with IABP for the treatment of critically Ill adult heart failure patients. Heart Lung Circ. 2014;23:363–8.

    PubMed  Google Scholar 

  33. Cheng JM, den Uil CA, Hoeks SE, van der Ent M, Jewbali LSD, van Domburg RT, et al. Percutaneous left ventricular assist devices vs. intra-aortic balloon pump counterpulsation for treatment of cardiogenic shock: a meta-analysis of controlled trials. Eur Heart J. 2009;30:2102–8.

    PubMed  Google Scholar 

  34. Koch CG, Li L, Duncan AI, Mihaljevic T, Cosgrove DM, Loop FD, et al. Morbidity and mortality risk associated with red blood cell and blood-component transfusion in isolated coronary artery bypass grafting. Crit Care Med. 2006;34:1608–16.

    PubMed  Google Scholar 

  35. Murphy GJ, Reeves BC, Rogers CA, Rizvi SIA, Culliford L, Angelini GD. Increased mortality, postoperative morbidity, and cost after red blood cell transfusion in patients having cardiac surgery. Circulation. 2007;116:2544–52.

    PubMed  Google Scholar 

  36. Loforte A, Montalto A, Ranocchi F, Della Monica PL, Casali G, Lappa A, et al. Peripheral extracorporeal membrane oxygenation system as salvage treatment of patients with refractory cardiogenic shock: preliminary outcome evaluation. Artif Organs. 2012;36:E53–61.

    CAS  PubMed  Google Scholar 

  37. Haddad F, Couture P, Tousignant C, Denault AY. The right ventricle in cardiac surgery, a perioperative perspective: pathophysiology, clinical importance, and management. Anesth Analg. 2009;108:422–33.

    PubMed  Google Scholar 

  38. Drakos SG, Janicki L, Kfoury AG, Reid BB, Clayson S, et al. Risk factors predictive of right ventricular failure after left ventricular assist device implantation. Am J Cardiol. 2010;105:1030–5.

    PubMed  Google Scholar 

  39. John R, Lee S, Eckman P, Liao K. Right ventricular failure: a continuing problem in patients with left ventricular assist device support. J Cardiovasc Transl Res. 2010;3:604–11.

    PubMed  Google Scholar 

  40. Ochiai Y, McCarthy PM, Smedira NG, Banbury MK, Navia JL, Feng J, et al. Predictors of severe right ventricular failure after implantable left ventricular assist device insertion: analysis of 245 patients. Circulation. 2002;106:198–202.

    Google Scholar 

  41. Matthews JC, Koelling TM, Pagani FD, Aaronson KD. The right ventricular failure risk score a pre-operative tool for assessing the risk of right ventricular failure in left ventricular assist device candidates. J Am Coll Cardiol. 2008;51:2163–72.

    PubMed  Google Scholar 

  42. Grant AD, Smedira NG, Starling RC, Marwick TH. Independent and incremental role of quantitative right ventricular evaluation for the prediction of right ventricular failure after left ventricular assist device implantation. J Am Coll Cardiol. 2012;60:521–8.

    PubMed  Google Scholar 

  43. Torgersen C, Schmittinger CA, Wagner S, Ulmer H, Takala J, Jakob SM, et al. Hemodynamic variables and mortality in cardiogenic shock: a retrospective cohort study. Crit Care. 2009;13:R157.

    PubMed  PubMed Central  Google Scholar 

  44. Samuels LE, Kaufman MS, Morris RJ. Pharmacologic criteria for ventricular assist device insertion: experience with the Abiomed BVS 5000 system. J Card Surg. 1999;288–93.

Download references

Author information

Authors and Affiliations

Authors

Corresponding author

Correspondence to Philippe Gaudard.

Additional information

Competing interests

The authors declare that they have no competing interests.

Authors’ contributions

PG supervised all aspects of the study, performed statistical analysis, and was involved in drafting the manuscript. MM developed the database, participated in the design of the study, the statistical analysis, the interpretation and the draft of the manuscript. JE participated in the design and development of the database and was involved in revising the manuscript. NZ and GC participated in data acquisition and development of the database and helped to draft the manuscript. PR and BA were involved in all surgical aspects of the study and participated in revising the manuscript. PC was involved in the design of the study, the interpretation of data and in revising the manuscript. All authors read and approved the final manuscript.

Rights and permissions

Open Access This article is distributed under the terms of the Creative Commons Attribution 4.0 International License (http://creativecommons.org/licenses/by/4.0/), which permits unrestricted use, distribution, and reproduction in any medium, provided you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons license, and indicate if changes were made. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated.

Reprints and Permissions

About this article

Check for updates. Verify currency and authenticity via CrossMark

Cite this article

Gaudard, P., Mourad, M., Eliet, J. et al. Management and outcome of patients supported with Impella 5.0 for refractory cardiogenic shock. Crit Care 19, 363 (2015). https://doi.org/10.1186/s13054-015-1073-8

Download citation

  • Received:

  • Accepted:

  • Published:

  • DOI: https://doi.org/10.1186/s13054-015-1073-8

Keywords

  • Right Ventricular
  • Cardiogenic Shock
  • Intensive Care Unit Discharge
  • Right Ventricular Function
  • Mechanical Circulatory Support