- Research
- Open Access
- Published:
Efficacy of positive end-expiratory pressure titration after the alveolar recruitment manoeuvre in patients with acute respiratory distress syndrome
Critical Care volume 13, Article number: R22 (2009)
Abstract
Introduction
In acute respiratory distress syndrome (ARDS), adequate positive end-expiratory pressure (PEEP) may decrease ventilator-induced lung injury by minimising overinflation and cyclic recruitment-derecruitment of the lung. We evaluated whether setting the PEEP using decremental PEEP titration after an alveolar recruitment manoeuvre (ARM) affects the clinical outcome in patients with ARDS.
Methods
Fifty-seven patients with early ARDS were randomly assigned to a group given decremental PEEP titration following ARM or a table-based PEEP (control) group. PEEP and inspired fraction of oxygen (FiO2) in the control group were set according to the table-based combinations of FiO2 and PEEP of the ARDS network, by which we aimed to achieve a PEEP level compatible with an oxygenation target. In the decremental PEEP titration group, the oxygen saturation and static compliance were monitored as the patients performed the ARM along with the extended sigh method, which is designed to gradually apply and withdraw a high distending pressure over a prolonged period, and the decremental titration of PEEP.
Results
The baseline characteristics did not differ significantly between the control and decremental PEEP titration groups. Initial oxygenation improved more in the decremental PEEP titration group than in the control group. However, dynamic compliance, tidal volume and PEEP were similar in the two groups during the first week. The duration of use of paralysing or sedative agents, mechanical ventilation, stay in the intensive care unit and mortality at 28 days did not differ significantly between the decremental PEEP titration and control groups.
Conclusions
The daily decremental PEEP titration after ARM showed only initial oxygenation improvement compared with the table-based PEEP method. Respiratory mechanics and patient outcomes did not differ between the decremental PEEP titration and control groups.
Trial registration
ClinicalTrials.gov identifier: ISRCTN79027921.
Introduction
Two recent randomised controlled trials involving patients with acute lung injury (ALI) or acute respiratory distress syndrome (ARDS) demonstrated that mortality can be reduced significantly by setting a low tidal volume (VT) [1] and by setting both a low VT and adequate positive end-expiratory pressure (PEEP) levels titrated by pressure-volume curves [2, 3]. However, this strategy favours further lung collapse or derecruitment, especially when used with a high inspired fraction of oxygen (FiO2) [4]. The importance of opening the lung and keeping it open seems increasingly significant in this era of lung-protective ventilatory support because the use of a small VT for this strategy may worsen progressive lung collapse. Moreover, the alveolar and systemic inflammatory responses can be attenuated by minimising overinflation and cyclic recruitment–derecruitment of the lung by reducing VT and increasing PEEP in ARDS [5].
The best method of setting optimal PEEP after recruitment to prevent recollapse is still a matter of debate [6]. In the ALVEOLI study, mortality rates and the number of ventilator-free days did not differ significantly between the lower- and higher-PEEP study groups [7]. This result may have reflected the use of the same PEEP level (combination of FiO2 and PEEP) for the heterogeneous patient group and an inappropriately high PEEP in the nonrecruiters, which may have resulted in overdistension. Gattinoni and colleagues reported that the percentage of potentially recruitable lung tissue may be different in patients with ARDS and that the use of a higher PEEP level in patients with a lower percentage of potentially recruitable lung may be harmful [8]. These studies did not evaluate completely the effect of using individualised PEEP on the survival rate of patients with ARDS. The higher PEEP levels should be adjusted according to individual lung morphology. However, the strategies for setting PEEP using a plateau pressure or recruitment manoeuvre do not decrease mortality [9, 10].
We designed an individual PEEP strategy using decremental PEEP titration after an alveolar recruitment manoeuvre (ARM) for each patient with ARDS. An ARM is a way to standardise the history of lung volume [11]. The objective of this study was to evaluate whether setting the PEEP using decremental PEEP titration after ARM affects the oxygenation and outcome of patients in the early stage of ARDS compared with the table-based combinations of FiO2 and PEEP in the ARDS network [1].
Materials and methods
Study population
Fifty-seven consecutive patients (35 men and 22 women) admitted to the medical intensive care unit (MICU) of Asan Medical Center, in Seoul, Korea, who were diagnosed with ARDS of various aetiologies were enrolled in the study between July 2004 and September 2006. Patient selection for the study was based on the criteria of ARDS proposed by the American–European Consensus Conference on ARDS [12]: acute onset, presence of hypoxaemia (partial arterial pressure of oxygen (PaO2)/FiO2 (PF ratio) ≤ 200 mmHg regardless of the PEEP level), bilateral and diffuse opacities seen on frontal chest x-ray and absence of left ventricular failure with pulmonary arterial occluded pressure of 18 mmHg of less.
The study protocol was approved by the institutional board of the ethics committee and written informed consent was obtained from the patients' families.
Ventilator procedures
Patients were given ventilatory support primarily in the supine position and ventilated according to the ARDS network strategy. FiO2, PEEP and respiratory rate were set to achieve an arterial oxygen saturation (SaO2) of between 88 and 92% [2]. The target VT was 6 ml/kg of predicted body weight, with allowances of up to 8 ml/kg if the SaO2 were below 88% or less than 7.2 of arterial pH by severe hypercapnoea. The patients were randomly assigned with the use of a randomisation scheme to either the decremental PEEP titration group or the control (table-based PEEP setting) group (Figure 1).
Patient group
Table-based PEEP setting (control) group
The FiO2-PEEP strategy has been used in previous ARDS Network studies (Figure 2). PEEP and FiO2 were set according to the table of lower PEEP/higher FiO2 combinations (the lower PEEP strategy of the ALVEOLI study), with the goal of obtaining a lower PEEP level compatible with an oxygenation target.
Decremental PEEP titration group
The ARM was performed immediately after enrollment in the study and was applied once a day usually in the morning for one week. The ventilatory circuit was not disconnected after ARM to avoid lung derecruitment. The ARM was also repeated when the ventilatory circuit was disconnected (incidentally or for bronchoscope) or if FiO2 requirement was increasing again in the patient, who showed initial improvement of oxygenation by ARM. If the weaning trial was performed within one week, the ARM was stopped earlier than the usual schedule. During the ARM, all patients were sedated and paralysed by continuous infusion of midazolam-ketamine and vecuronium bromide. No changes were made in the doses of inotropic agents or fluid infusion during the ARM.
The ARM used the extended sigh method, which is designed to gradually apply and withdraw a high distending pressure over a prolonged period. It takes about 15 to 20 minutes for two cycles of ARM according to our protocol (Figure 2) [13]. We changed the ventilatory mode from pressure-controlled mode to the volume-controlled mode during ARM. The distending pressure was determined by the delivered VT, PEEP increment and the pause time (0.5 seconds). During the ARM, PEEP was added from baseline to 15, 20 and 25 cmH2O sequentially (every 30 seconds from the baseline PEEP until 25 cmH2O). The VT was decreased by 25% from the baseline VT during the incremental PEEP trial phase, and then returned to baseline levels during the decremental PEEP trial phase. Therefore, the distending pressures were changed depending on the patient's VT and lung mechanics. However, we did not allow the peak airway pressure to go above 55 cmH2O during the ARM. The decremental PEEP titration at the second cycle of ARM was performed with progressive decreases in PEEP in steps of 1 cmH2O every 30 seconds from 20 cmH2O while continuously monitoring saturation and static compliance. The decrease in PEEP was continued until a decrease of more than 2% of saturation from the previous SaO2 and drop of static compliance was identified. This PEEP level was considered the alveolar collapsing pressure and the optimal PEEP after the ARM was set 2 cmH2O above this pressure. No patients showed significant arrhythmia or gross barotraumas of any type during ARM.
Rescue therapy
If the level of inspired oxygen was not decreased to 0.6 or oxygenation improvement was not achieved after PEEP readjustment in all patients, rescue therapies such as prone position or nitric oxide inhalation were performed.
Outcome measures and data collection
The primary end point was improvement in oxygenation (improvement of PaO2/FiO2 (PF) ratio). The secondary end points included respiratory mechanics (PEEP and dynamic compliance), ICU stay, duration of sedatives and paralysing agents and patient outcomes (28-day mortality, 60-day mortality, duration of mechanical ventilation). Responders were defined by a 20% improvement in the PF ratio on day 1 compared with day 0 (baseline) after PEEP adjustment [14].
Respiratory mechanics
Airway pressure and flow were monitored continuously. VT, dynamic compliance (in ml/cmH2O), and peak airway, mean airway and minute ventilation were recorded at 30 minutes after the change of PEEP level in both groups. Because pressure control mode was the main ventilatory strategy, we monitored and compared the dynamic compliance and peak airway pressure in both group.
Haemodynamics and gas exchange
We collected arterial blood samples to measure partial pressure of oxygen (PO2), partial pressure of carbon dioxide (PCO2), pH, and SaO2 30 minutes after reapplication of PEEP. Haemodynamic variables monitored included heart rate and systolic, diastolic and mean systemic arterial pressure.
Data analysis
Sample size calculation showed that 36 patients per group would provide 80% power at a two-sided α level of 0.05 to detect a 20% difference in the improvement of oxygenation (improvement of PF ratio).
All data were analysed using SPSS for Windows (version 11.0; SPSS Inc., Chicago, IL, USA). Statistical analyses were based on the intention-to-treat principle and involved all patients who had undergone randomisation. All values are expressed as the mean ± standard error of the mean or as the number and percentage of patients. Probability of mortality and differences between the groups were compared using a log-rank test. The chi-squared test or Fisher's exact test was used to compare categorical data, and Student's t test or the Mann-Whitney U test was used to compare continuous data. Significance was defined as p < 0.05.
Results
Characteristics of the patients
We enrolled 61 patients in the study, 30 of whom were randomly assigned to the control (table-based PEEP setting) group and 31 to the decremental PEEP titration group. Fifty-seven patients completed the study (27 in the control group and 30 in the decremental PEEP titration group) and contributed data for the analyses. Most baseline characteristics were similar in the two study groups (Table 1). The most common cause of ARDS was pneumonia. The initial ventilatory setting and severity index did not differ between the control and decremental PEEP titration groups.
Respiratory mechanics and oxygenation
Figure 3 shows the ventilatory setting and respiratory variables at baseline and follow-up during the first week of treatment. The peak pressure was significantly higher on days 3 and 5 in the decremental PEEP titration group than in the control group, and the mean pressure was significantly higher in the decremental PEEP titration group because of the higher PEEP during the first week. Dynamic compliance and VT during the first week were similar in the two groups. Oxygenation improved compared with the baseline PF ratio in both groups (Figure 4). Initial oxygenation improved more in the decremental PEEP titration group than in the control group. The partial arterial pressure of oxygen (PaCO2) level was significantly higher in the decremental PEEP titration group than in the control group on day 1, suggesting increased dead space ventilation. However, the improvement of oxygenation and PaCO2 level were not different between the two groups during follow-up.
Respiratory values during the first week of treatment. The closed circles indicate the control group and the open circles denote the decremental positive end-expiratory pressure (PEEP) titration group. Values are expressed as the mean ± standard error of the mean (bars). *p < 0.05 between the control and decremental PEEP titration groups; †p < 0.05 compared with day 0 in the control group; ‡p < 0.05 compared with day 0 in the decremental PEEP titration group.
Oxygenation changes and PaCO2 levels during the first week of treatment. All patients showed improved oxygenation during treatment. The level of partial arterial pressure of carbon dioxide (PaCO2) at day 1 was increased significantly in the decremental positive end-expiratory pressure (PEEP) titration group than the control group. The black bars indicate the control group and gray bars denote the decremental PEEP titration group. Values are expressed as the mean ± standard error of the mean (bars). *p < 0.05 between the control and decremental PEEP titration groups, †p < 0.05 compared with day 0 in the control group, ‡p < 0.05 compared with day 0 in the decremental PEEP titration group.
Clinical outcomes
The overall mortality at 28 days was 37%. Mortality at 28 days was 33% in the control group and 40% in the decremental PEEP titration group (Table 2). Using Cox regression for 28-day mortality, the survival rate in the decremental PEEP titration group was not different (p = 0.725; hazard ratio = 1.168; 95% confidence interval = 0.493 to 2.768). However, 60-day mortality was significantly increased to 55.6% (p = 0.031) compared with 28-day mortality (33.3%) in the control group only.
Eight patients (30%) in the control group and six patients (20%) in the decremental PEEP titration group had performed the weaning trial within a week of enrollment. The incidence of barotrauma was similar in both groups. The durations of mechanical ventilation, ICU stay and use of paralysing or sedative agents did not differ between the groups (Table 2).
Discussion
As suggested by Lachmann more than 10 years ago [15], "open up the lung and keep the lung open" appears advantageous to recruit the lungs of patients with ARDS and to prevent subsequent lung derecruitment [12]. Clinical outcomes were similar regardless of whether lower or higher PEEP levels were used in the ALVEOLI trial [7]. The problem of the ALVEOLI trial was the higher PEEP strategy, which was table-based like the one on the current study, matched on the oxygenation target regardless of any patient-related variable. A reasonable approach to determining the appropriate level of PEEP requires maintaining PEEP-induced reopening of atelectatic areas and avoiding PEEP-induced lung overinflation. Higher than traditional PEEP levels together with lung-recruiting maneuvers seems to be a way to find an appropriate PEEP level [16].
Experimental data suggest that the effects of ARM on alveolar recruitment are transient if the preceding PEEP levels are maintained after the manoeuvre [14]. Once the alveoli have been recruited, higher PEEP levels are required to keep them aerated [17]. The right level of PEEP as an anti-derecruiting force is important in preserving the effect of the ARM. We previously reported that a sufficient level of PEEP after ARM is important as an anti-decruitment strategy [18].
We hypothesised that the decremental PEEP trial would be an appropriate method to establish the PEEP level after ARM at bedside. Because higher PEEP and the ARM performed in patients with mild lung injury may have fewer benefits and more adverse effects, we only focused on ARDS patients to address our assumption. We found that daily decremental PEEP titration after ARM showed only initial oxygenation improvement compared with the table-based PEEP method and did not improve the respiratory mechanics within a week. We performed the ARM and PEEP titration daily during the first week in the decremental PEEP titration group. However, no significant differences were observed in the 28-day mortality, ICU stay and 60-day mortality. Although the responder rate was higher in the decremental PEEP titration group than in the control group, the earlier improvement in oxygenation was not associated with increased survival rate.
Our finding showed the 60-day mortality in the control group was significantly increased compared with the 28-day mortality. We could not explain whether this finding was associated with the protective effect of ARM to the ventilator-induced lung injury. To add any relevant information, further study will be needed for the biomarkers such as proinflammatory cytokines.
The lower dynamic compliance in the decremental PEEP titration group was an unexpected result, although there was no significant difference between the two groups. We speculated that a higher peak airway pressure might affect the lower dynamic compliance in the decremental PEEP titration group than the control. A reason for this result might be related to the subjects' characteristics. Most of the patients with ARDS had pneumonia, which did not respond well to the applied PEEP [19].
We included mostly pulmonary ARDS with severe underlying diseases, which would be a common phenomenon in a university hospital MICU. That may be the reason why mortality related to progressive respiratory failure was higher in our study group. The mortality of patients with pneumonia was 43.3% and the mortality of patients without pneumonia was 29.6%. The baseline lung conditions of these patients was a higher proportion of refractory consolidation to the distending pressure which may also have influenced the PF ratio of less than 250 and PEEP levels less than 15 cmH2O after ARM. Talmor and colleagues reported that a low lung distending pressure was applied in the patients with a stiff chest wall [20]. We did not measure the chest-wall mechanics or the recruitable lung, so we could not address this possibility as a cause of low PF ratio.
One limitation of our study is that we could not evaluate whether recruitment manoeuvre maximised the alveolar collapse. If the recruitment effect of our ARM was not sufficient, the effect of a decremental PEEP trial might be insufficient. Borges and colleagues reported that a peak airway pressure of more than 60 cmH2O could recruit the collapsed lung [21]. Another limitation is that our indicators of oxygen saturation and static compliance to find an appropriate PEEP level might be insensitive to detect the collapsing pressure of the lung. Finally, small sample size in a single centre limits the power of the study outcomes.
Conclusion
The daily decremental PEEP titration after ARM did not show the persistent improvement of oxygenation, the respiratory mechanics and the mortality rate of patients with ARDS compared with the table-based PEEP setting. Further investigations are needed to find the correct level of PEEP in ARDS with reference to chest wall compliance and alveoli mechanics.
Key messages
-
The daily decremental PEEP titration after ARM method did not show a persistent improvement of oxygenation.
-
Respiratory mechanics such as dynamic compliance, VT and PEEP were not significantly different between the daily decremental PEEP titration after ARM group and the table-based PEEP setting group.
-
The daily decremental PEEP titration after ARM without the reference of lung mechanics did not reduce the dependence of ventilator or mortality rate of patients with ARDS compared with the table-based PEEP setting.
Abbreviations
- ALI:
-
acute lung injury
- ARDS:
-
acute respiratory distress syndrome
- ARM:
-
alveolar recruitment manoeuvre
- FiO2:
-
inspired fracture of oxygen
- MICU:
-
medical intensive care unit
- PEEP:
-
positive end-expiratory pressure
- PaCO2:
-
partial arterial pressure of carbon dioxide
- PaO2:
-
partial arterial pressure of oxygen
- PCO2:
-
partial pressure of carbon dioxide
- PO2:
-
partial pressure of oxygen
- SaO2:
-
arterial oxygen saturation
- VT:
-
tidal volume.
References
- 1.
Ventilation with lower tidal volumes as compared with traditional tidal volumes for acute lung injury and the acute respiratory distress syndrome. The Acute Respiratory Distress Syndrome Network N Engl J Med 2000, 342: 1301-1308. 10.1056/NEJM200005043421801
- 2.
Amato MB, Barbas CS, Medeiros DM, Magaldi RB, Schettino GP, Lorenzi-Filho G, Kairalla RA, Deheinzelin D, Munoz C, Oliveira R, Takagaki TY, Carvalho CR: Effect of a protective-ventilation strategy on mortality in the acute respiratory distress syndrome. N Engl J Med 1998, 338: 347-354. 10.1056/NEJM199802053380602
- 3.
Villar J, Kacmarek RM, Perez-Mendez L, Aguirre-Jaime A: A high positive end-expiratory pressure, low tidal volume ventilatory strategy improves outcome in persistent acute respiratory distress syndrome: a randomized, controlled trial. Crit Care Med 2006, 34: 1311-1318. 10.1097/01.CCM.0000215598.84885.01
- 4.
Santos C, Ferrer M, Roca J, Torres A, Hernandez C, Rodriguez-Roisin R: Pulmonary gas exchange response to oxygen breathing in acute lung injury. Am J Respir Crit Care Med 2000, 161: 26-31.
- 5.
Ranieri VM, Suter PM, Tortorella C, De Tullio R, Dayer JM, Brienza A, Bruno F, Slutsky AS: Effect of mechanical ventilation on inflammatory mediators in patients with acute respiratory distress syndrome: a randomized controlled trial. JAMA 1999, 282: 54-61. 10.1001/jama.282.1.54
- 6.
Rouby JJ, Lu Q, Goldstein I: Selecting the right level of positive end-expiratory pressure in patients with acute respiratory distress syndrome. Am J Respir Crit Care Med 2002, 165: 1182-1186.
- 7.
Brower RG, Lanken PN, MacIntyre N, Matthay MA, Morris A, Ancukiewicz M, Schoenfeld D, Thompson BT: Higher versus lower positive end-expiratory pressures in patients with the acute respiratory distress syndrome. N Engl J Med 2004, 351: 327-336. 10.1056/NEJMoa032193
- 8.
Gattinoni L, Caironi P, Cressoni M, Chiumello D, Ranieri VM, Quintel M, Russo S, Patroniti N, Cornejo R, Bugedo G: Lung recruitment in patients with the acute respiratory distress syndrome. N Engl J Med 2006, 354: 1775-1786. 10.1056/NEJMoa052052
- 9.
Mercat A, Richard JC, Vielle B, Jaber S, Osman D, Diehl JL, Lefrant JY, Prat G, Richecoeur J, Nieszkowska A, Gervais C, Baudot J, Bouadma L, Brochard L: Positive end-expiratory pressure setting in adults with acute lung injury and acute respiratory distress syndrome: a randomized controlled trial. JAMA 2008, 299: 646-655. 10.1001/jama.299.6.646
- 10.
Meade MO, Cook DJ, Guyatt GH, Slutsky AS, Arabi YM, Cooper DJ, Davies AR, Hand LE, Zhou Q, Thabane L, Austin P, Lapinsky S, Baxter A, Russell J, Skrobik Y, Ronco JJ, Stewart TE: Ventilation strategy using low tidal volumes, recruitment maneuvers, and high positive end-expiratory pressure for acute lung injury and acute respiratory distress syndrome: a randomized controlled trial. JAMA 2008, 299: 637-645. 10.1001/jama.299.6.637
- 11.
Nishida T, Suchodolski K, Schettino GP, Sedeek K, Takeuch M, Kacmarek RM: Peak volume history and peak pressure-volume curve pressures independently affect the shape of the pressure-volume curve of the respiratory system. Crit Care Med 2004, 32: 1358-1364. 10.1097/01.CCM.0000128573.28173.2E
- 12.
Artigas A, Bernard GR, Carlet J, Dreyfuss D, Gattinoni L, Hudson L, Lamy M, Marini JJ, Matthay MA, Pinsky MR, Spragg R, Suter PM: The American-European Consensus Conference on ARDS, part 2: ventilatory, pharmacologic, supportive therapy, study design strategies, and issues related to recovery and remodeling. Acute respiratory distress syndrome. Am J Respir Crit Care Med 1998, 157: 1332-1347.
- 13.
Lim CM, Koh Y, Park W, Chin JY, Shim TS, Lee SD, Kim WS, Kim DS, Kim WD: Mechanistic scheme and effect of "extended sigh" as a recruitment maneuver in patients with acute respiratory distress syndrome: a preliminary study. Crit Care Med 2001, 29: 1255-1260. 10.1097/00003246-200106000-00037
- 14.
Grasso S, Mascia L, Del Turco M, Malacarne P, Giunta F, Brochard L, Slutsky AS, Marco Ranieri V: Effects of recruiting maneuvers in patients with acute respiratory distress syndrome ventilated with protective ventilatory strategy. Anesthesiology 2002, 96: 795-802. 10.1097/00000542-200204000-00005
- 15.
Lachmann B: Open up the lung and keep the lung open. Intensive Care Med 1992, 18: 319-321. 10.1007/BF01694358
- 16.
Medoff BD, Harris RS, Kesselman H, Venegas J, Amato MB, Hess D: Use of recruitment maneuvers and high-positive end-expiratory pressure in a patient with acute respiratory distress syndrome. Crit Care Med 2000, 28: 1210-1216. 10.1097/00003246-200004000-00051
- 17.
Halter JM, Steinberg JM, Schiller HJ, DaSilva M, Gatto LA, Landas S, Nieman GF: Positive end-expiratory pressure after a recruitment maneuver prevents both alveolar collapse and recruitment/derecruitment. Am J Respir Crit Care Med 2003, 167: 1620-1626. 10.1164/rccm.200205-435OC
- 18.
Lim CM, Jung H, Koh Y, Lee JS, Shim TS, Lee SD, Kim WS, Kim DS, Kim WD: Effect of alveolar recruitment maneuver in early acute respiratory distress syndrome according to antiderecruitment strategy, etiological category of diffuse lung injury, and body position of the patient. Crit Care Med 2003, 31: 411-418. 10.1097/01.CCM.0000048631.88155.39
- 19.
Gattinoni L, Pelosi P, Suter PM, Pedoto A, Vercesi P, Lissoni A: Acute respiratory distress syndrome caused by pulmonary and extrapulmonary disease. Different syndromes? Am J Respir Crit Care Med 1998, 158: 3-11.
- 20.
Talmor D, Sarge T, Malhotra A, O'Donnell CR, Ritz R, Lisbon A, Novack V, Loring SH: Mechanical ventilation guided by esophageal pressure in acute lung injury. N Engl J Med 2008, 359: 2095-2104. 10.1056/NEJMoa0708638
- 21.
Borges JB, Okamoto VN, Matos GF, Caramez MP, Arantes PR, Barros F, Souza CE, Victorino JA, Kacmarek RM, Barbas CS, Carvalho CR, Amato MB: Reversibility of lung collapse and hypoxemia in early acute respiratory distress syndrome. Am J Respir Crit Care Med 2006, 174: 268-278. 10.1164/rccm.200506-976OC
Author information
Affiliations
Corresponding author
Additional information
Competing interests
The authors declare that they have no competing interests.
Authors' contributions
JWH recruited patients, analysed the data and wrote the manuscript. HJ and HSC helped to recruit patients and analysed the data. SBH and CML recruited patients and interpreted the data. YK designed the study, interpreted the data and wrote the paper. All authors read and approved the final manuscript.
Authors’ original submitted files for images
Below are the links to the authors’ original submitted files for images.
Rights and permissions
About this article
Cite this article
Huh, J.W., Jung, H., Choi, H.S. et al. Efficacy of positive end-expiratory pressure titration after the alveolar recruitment manoeuvre in patients with acute respiratory distress syndrome. Crit Care 13, R22 (2009). https://doi.org/10.1186/cc7725
Received:
Accepted:
Published:
Keywords
- Acute Respiratory Distress Syndrome
- Respiratory Mechanic
- Peep Level
- Dynamic Compliance
- High Peep