Skip to main content

Short-term effects of positive end-expiratory pressure on breathing pattern: an interventional study in adult intensive care patients

Abstract

Introduction

Positive end-expiratory pressure (PEEP) is used in mechanically ventilated patients to increase pulmonary volume and improve gas exchange. However, in clinical practice and with respect to adult, ventilator-dependent patients, little is known about the short-term effects of PEEP on breathing patterns.

Methods

In 30 tracheally intubated, spontaneously breathing patients, we sequentially applied PEEP to the trachea at 0, 5 and 10 cmH2O, and then again at 5 cmH2O for 30 s each, using the automatic tube compensation mode.

Results

Increases in PEEP were strongly associated with drops in minute ventilation (P < 0.0001) and respiratory rate (P < 0.0001). For respiratory rate, a 1 cmH2O change in PEEP in either direction resulted in a change in rate of 0.4 breaths/min. The effects were exclusively due to changes in expiratory time. Effects began to manifest during the first breath and became fully established in the second breath for each PEEP level. Identical responses were found when PEEP levels were applied for 10 or 60 s. Post hoc analysis revealed a similar but stronger response in patients with impaired respiratory system compliance.

Conclusion

In tracheally intubated, spontaneously breathing adult patients, the level of PEEP significantly influences the resting short-term breathing pattern by selectively affecting expiratory time. These findings are best explained by the Hering–Breuer inflation/deflation reflex.

Introduction

Pulmonary stretch receptors affect the resting respiratory pattern by vagal afferents. Increased stretch receptor activity shortens the duration of inspiration in animals [1], human newborns, and children [29]. This reflex is well known as the Hering–Breuer inflation reflex [10]. Increased stretch receptor activity also prolongs expiration, and maintained inflations – if sufficiently large – can produce apnoea for a considerable period of time. In humans the inflation/deflation reflex is substantially weaker than in most animal species [1, 8]. In adult humans, the inflation/deflation reflex has been found to become apparent if the inflation volume exceeds a critical threshold of about 1 l [3, 9, 11]. Recent work revealed that the inflation/deflation reflex is also operative in normal tidal breathing if subject is sleeping [12, 13] or under slight sedation [14]. Moreover, Tryfon and coworkers [14] demonstrated that raising the level of continuous positive airway pressure (or positive end-expiratory pressure [PEEP]) significantly prolongs expiration in parallel with a PEEP-related increase in functional residual capacity. However, the findings of those studies [1214] became apparent only after an inspiratory hold manoeuvre and are therefore beyond the limits of physiological or common clinical conditions.

Our aim in the present study was to investigate whether these effects also become apparent during standard clinical conditions. To this end, we repeatedly applied a pattern of different levels of PEEP in a heterogeneous population of tracheally intubated patients during unsupported spontaneous breathing. Using this approach we were able to demonstrate that increased PEEP was strongly associated with prolongation of the expiratory time (resulting in a fall in both respiratory rate and minute ventilation [VE]), whereas tidal volume (VT) and inspiratory time were not significantly affected.

Materials and methods

Patients

Thirty patients were investigated during weaning from mechanical ventilation. The characteristics of the patients and their underlying diseases are summarized in Table 1. Sedation therapy had been discontinued in 21 (70%) patients 6 ± 7 hours before the start of the study. At entrance into the study, all patients were either awake or easy to arouse, corresponding to a Ramsey sedation score of 2 or 3 [15]. Patients were breathing spontaneously in the pressure support ventilation (PSV) mode with an inspiratory pressure assist of 6–12 cmH2O above a PEEP of 5–10 cmH2O. For a patient to be enrolled in the study, the fractional inspired oxygen (FiO2) required to maintain their arterial oxygen tension and arterial oxygen saturation above 10 kPa (75 mmHg) and 92%, respectively, had to be 50% or less. Furthermore, the patients had to be able to maintain blood gas values within normal ranges, to exhibit no severe coughing or repeated swallowing, or obvious signs of discomfort, and to have stable cardiovascular and respiratory conditions (i.e. absence of systematic changes to ventilatory pattern, FiO2 or PEEP requirements, or haemodynamic conditions). Of these patients, 19 (63%) could be successfully extubated within 1–24 hours after termination of the study. The remaining 11 patients (37 %) had to be kept on the ventilator for a mean of 3 ± 2 days for prolonged weaning from mechanical ventilation for various reasons.

Table 1 Patient characteristics

Study design

The study protocol was approved by the ethics committee of our institution, and informed consent was obtained either from the patient (in the case of elective surgery) or from the patient's next of kin. After study inclusion criteria had been fulfilled, patients were connected to a modified study ventilator [16], in which the level of PEEP was set to 5 cmH2O. If necessary, FiO2 was adjusted to maintain oxygenation within comfortable ranges (i.e. arterial oxygen tension and arterial oxygen sturation above 11 kPa [82.5 mmHg] and 94%, respectively). After 20-min period to stabilize the temperature and humidity of the system and to allow the patient to become accustomed to the ventilator, the ventilatory mode was switched from conventional PSV to the automatic tube compensation mode [16, 17]. We used automatic tube compensation mode (i.e. tracheal continuous positive airway pressure) instead of conventional PSV mode in order to eliminate the influence of flow-dependent pressure drop across the endotracheal tube (ETT) [17, 18].

The study ventilator was driven by an external control unit, which – in accordance with the study protocol – automatically changes the PEEP level in synchrony with the patient's breathing pattern (i.e. upward steps in PEEP during inspiration and downward steps during expiration). Even though this helps to minimize interference with the patient's breathing pattern, the duration of expiration of the first breath after PEEP adjustment was shortened and prolonged, respectively, by upward steps and downward steps in PEEP. Consequently, these breaths had to be rejected from our analyses.

Instead of airway pressure, the tracheal pressure was the target for PEEP adjustment. To this end, the tracheal pressure was continuously calculated [19] and repeated spot-check measurements between investigations guaranteed for quality of the calculation procedure [20]. The study protocol includes three levels of PEEP that were applied in the following sequence: 0, 5 and 10 cmH2O, and the 5 cmH2O again. Each level of PEEP was applied for 30 s (Fig. 1). To minimize time-dependent biases, the PEEP sequence was applied five times in a row. Additionally, we applied the study protocol for intervals of 10 s (n = 25) and 60 s (n = 18).

Figure 1
figure 1

Time course of tracheal pressure during steps in PEEP. The examination was performed while the patient was breathing spontaneously at (tracheal) continuous positive airway pressure by means of the automatic tube compensation mode. Starting from 5 cmH2O, the level of PEEP was changed to 0, 5, 10 and (again) 5 cmH2O for durations of 30 s each for five consecutive runs. Note that the changes in PEEP level were in synchrony with the patient's breathing pattern (i.e. upward steps during inspiration and downward steps during expiration). PEEP, positive end-expiratory pressure; Ptrach, tracheal pressure.

Measurements

The flow (V') was measured using a heated Fleisch No. 2 pneumotachograph (Metabo, Epalinges, Switzerland), which was placed at the proximal end of the ETT. Airway pressure was measured between the pneumotachograph and the outer end of the ETT. Spot-check measurement of tracheal pressure were taken by introducing a pressure measuring catheter into the trachea via the lumen of the ETT, as described in detail elsewhere [20]. The pressure transducers used to measure airway and tracheal pressures (32NA-005D; ICsensor, Milpitas, CA, USA) and the differential pressure transducer for measuring V' (CPS 10; Hoffrichter, Schwerin, Germany) were placed a small distance from the patient (20 cm) to achieve a good signal quality and short response time. Expiratory carbon dioxide was measured using a device (CO2 Analyzer 930; Siemens-Elema, Solna, Sweden) that was previously calibrated with a carbon dioxide concentration of 8.4%. Measured signals were digitized with 12-bit precision and stored at a rate of 100 Hz in a personal computer for further analyses. The mechanics of the respiratory system (i.e. compliance, resistance and intrinsic PEEP) were calculated from tracings originating from a preceding period of controlled mechanical ventilation [21, 22], in which resistance is depicted as pure airway resistance (i.e. without the resistance of the ETT).

Analysis

Stored data were analyzed on a breath-by-breath basis after having rejected the breaths that were affected by transitions in PEEP level. We determined inspiratory time, expiratory time and respiratory rate by means of the flow signal, in which a combined criterion of flow and expiratory carbon dioxide concentration allowed us to differentiate between inspiration and expiration accurately [23]. VT was calculated by numerical integration of V'. VE was calculated by multiplying respiratory rate by VT. In 10 out of 30 patients, the exhaled carbon dioxide volume per minute was derived from the additive compound of flow and carbon dioxide samples during expiration. For subsequent statistical analysis, data for corresponding PEEP levels and identical time cues were averaged.

In a post hoc analysis we investigated the PEEP-related effect on breathing pattern in patients without (control group; n = 17) and with impaired lung mechanics (impaired group; n = 13). Allocation was roughly based on respiratory system compliance above or below 50 ml/cmH2O (Table 2). Because at the time of investigation pure airway resistance (i.e. without ETT resistance) was below 4 cmH2O·s/l in each patient, allocation according to airway resistance was not feasible.

Table 2 Baseline characteristics: control versus impaired respiratory system mechanics

Statistical analysis was performed using SYSTAT, version 5.2 (L. Wilkinson, M.A. Hill, E. Vang, Evanston, IL, USA). Differences in all parameters of interest were assessed by an analysis of variance suitable for repeated measures with four repeated within factors (i.e. the PEEP levels) and either no between factor (main analysis) or one between factor (group) in the post hoc analysis. Significance was expressed as the arithmetic mean of Greenhouse–Geisser's and Huynh–Feldt's adjusted P values. If the overall model showed significant results, then changes between PEEP levels were calculated. For baseline group differences in the post hoc analysis, the Kruskal–Wallis test was used to compare continuous variables and the χ2 test was used to compare categorical data. For all tests a two-sided α level of P < 0.025 was considered statistically significant. All data are presented as mean ± standard deviation, unless otherwise stated.

Results

For the investigation with 30 s intervals, a total of 175 ± 88 breaths per patient were eligible for statistical analysis. In each patient the increase in PEEP from 0 to 10 cmH2O was associated with a fall in VE from on average 11.6 ± 3.0 l/min to 10.0 ± 2.0 l/min (P < 0.0001). Whereas VT was unaffected by the increase in PEEP (545 ± 184 ml versus 550 ± 163 ml at PEEP 0 and 10 cmH2O, respectively; P = 0.571), the fall in VE was due to a decrease in respiratory rate from 23.6 ± 9.6 breaths/min at zero PEEP to 19.9 ± 7.3 breaths/min at 10 cmH2O PEEP (P = 0.001). The decrease in respiratory rate was due to a significant increase in expiratory time (from 2111 ± 893 ms at zero PEEP to 2599 ± 1047 ms at 10 cmH2O PEEP; P < 0.0001), whereas inspiratory time remained unaffected (869 ± 321 ms and 880 ± 312 ms at PEEP 0 and 10 cmH2O, respectively; P = 0.116). Expiratory carbon dioxide volume significantly decreased with increasing levels of PEEP (P < 0.01), whereas the end-expiratory carbon dioxide did not differ significantly.

For the variables of interest, the differences between all levels of PEEP are shown in Table 3; Fig. 2 shows the corresponding percentage changes. Figures 3 and 4 show the PEEP-related effects on VE and expiratory time in individual patients. An identical pattern was found when the PEEP pattern was applied for intervals of 10 s or 60 s (Fig. 5). Figure 6 shows changes in breathing pattern in the first, second and third breaths after steps of PEEP in either direction. Changes in breathing pattern had begun to manifest during the first breath and became fully established in the second breath after both upward and downward steps in PEEP.

Table 3 Changes in breathing pattern produced by gradual changes in PEEP
Figure 2
figure 2

Short-term effects of PEEP on breathing pattern. The increase in PEEP is paralleled by an increase in expiratory time with a concomitant fall in both respiratory rate and minute ventilation. The decrease in PEEP has an opposite effect. Tidal volume and inspiratory time were not significantly affected by changes in PEEP in either direction. Results are outlined as averaged percentage changes from zero PEEP and are expressed as mean ± 1 standard error of the mean (SEM). *P < 0.025, versus zero PEEP. PEEP, positive end-expiratory pressure.

Figure 3
figure 3

Short-term effects of PEEP on VE in the individual patient. Results are from the investigation with PEEP steps of 30 s duration, and each coloured line shows findings for a different patient. Although VE is dispersed over a wide range, its behaviour was similar between patients (i.e. VE decreased with increasing levels of PEEP, and vice versa). PEEP, positive end-expiratory pressure; VE, minute ventilation.

Figure 4
figure 4

Short-term effects of PEEP on expiratory time in the individual patient. Results are from the investigation with PEEP steps of 30 s duration, and each coloured line shows findings for a different patient. Although Tex is dispersed over a wide range, its behaviour was similar between patients (i.e. Tex increased with the increasing levels of PEEP, and vice versa). PEEP, positive end-expiratory pressure; Tex, expiratory time.

Figure 5
figure 5

Effects on breathing pattern brought on by steps in PEEP of different duration. Effects are shown on VE, which behaved similarly whether the PEEP steps were applied for 10, 30, or 60 s. *P < 0.025, versus zero PEEP. PEEP, positive end-expiratory pressure; VE, minute ventilation.

Figure 6
figure 6

Changes of breathing pattern in the first, second and third breath after steps of PEEP. Results are from the investigation with PEEP steps of 30 s duration. The shaded areas represent averaged values of all breath for the corresponding PEEP level; open bars indicate the first, second, and third breath after changes of PEEP. Values are expressed as means ± 1 standard error of the mean (SEM). Note that changes of breathing pattern were beginning to manifest within the first breath and became fully established in the second breath after both upward and downward steps in PEEP. *P < 0.025, versus mean values within the corresponding PEEP level. BTPS, body temperature pressure, (water damp) saturated; PEEP, positive end-expiratory pressure; rr, respiratory rate; Tex, expiratory time; Tin, inspiratory time; VE, minute ventilation.

Post hoc analysis

In patients with decreased respiratory system compliance, respiratory rate and VE were significantly higher and inspiratory time, expiratory time and VT were significantly smaller as compared with the corresponding parameters in the control group. Irrespective of these group differences, the PEEP-related effect on breathing pattern was the same in both groups (i.e. expiratory time increased, VE and respiratory rate decreased, and inspiratory time and VT remained unaffected by increases in the level of PEEP, and vice versa). Effects were more pronounced in patients with impaired respiratory system compliance than in the control group (P < 0.025). In both groups, the changes became fully established at the latest within the second breath after a change in PEEP.

Discussion

The findings of this clinical study, conducted in a heterogeneous population of adult intensive care patients, indicate that the level of PEEP significantly influences resting short-term breathing patterns by selectively affecting the duration of expiration. Thus, a reduction in PEEP is paralleled by an increase in respiratory rate and subsequently in VE, and vice versa. According to our findings, the magnitude of the effect is about 0.4 breaths/min per 1 cmH2O change in PEEP in either direction, and so it is about 10 times smaller than the effect found in anaesthetized animals [24]. Our findings are in accordance with the results of other studies in adult humans at normal tidal breathing [1214], in which the findings were attributed to the Hering–Breuer inflation/deflation reflex [14]. In those studies, however, the effect became apparent only with an inspiratory hold manoeuvre. In contrast, we were able to show this effect also at normal tidal breathing under quite common clinical conditions (i.e. without any dedicated respiratory manoeuvre). Furthermore, we were also able to demonstrate that the PEEP-related effect had already begun to manifest during the first breath and became fully established in the second breath after adjustment to PEEP level. Consistent with these findings, we could not find any substantial difference when the study protocol was applied with PEEP durations as short as 10 s and up to 60 s.

The PEEP-related effect upon breathing pattern found in our heterogeneous study population was not only significant but also substantial. However, whether these findings are of clinical importance remains unclear. On the one hand, this is because in our study we focused on short-term rather than on long-term effects. From a theoretical point of view, the short-term responses seen here might simply be offset by a subsequent shift in blood carbon dioxide concentration beyond the time frame of 60 s. Such an hypothesis is strongly supported by the slight but significant decrease in exhaled carbon dioxide volume with increasing levels of PEEP (as a result of the PEEP-related drop in VE).

To investigate whether the PEEP-related short-term effect is only transient, we applied our study protocol to 10 of the 28 patients for durations of 180 and 300 s per level of PEEP. Unfortunately, the obtained data were inconsistent for the vast majority of these patients. This was due to the increased rate of artifacts (mainly due to restlessness during awakening) associated with the prolonged duration of investigation. Consequently, because our study was not designed to be a long-term investigation, we are unable to draw any conclusions on whether the observed short-term responses are preserved over time or whether they are offset by slowly adapting reflexes or behavioural responses over a longer period of observation. On the other hand, the PEEP-related short-term effect is sufficiently substantial that it could be detected by attentive clinicians, and so it could be used to adapt ventilatory settings for weaning from mechanical ventilation. Furthermore, short-term effects are of the utmost importance with respect to the automated weaning procedures and closed-loop ventilation strategies that now are increasingly being applied worldwide [25].

The question then arises as to the mechanism by which changes in PEEP can affect the breathing pattern, as found in this study. Based on findings in animals and humans, it is likely that the PEEP-related effect is best explained by the Hering–Breuer inflation/deflation reflex [3, 8, 10, 2628]. However, other mechanisms are also possible. First, the effect might be due to behavioural responses. At study entrance, all patients were either awake or easy to arouse because sedation had been discontinued 6 ± 7 hours before the start of the study in most patients and continued at a rather slight level in a few patients. Nevertheless, the PEEP-related responses were quite uniform; for example, they became fully established within the first two breaths after PEEP adjustment in virtually all patients. Furthermore, we did not find any difference in the PEEP-related effect between patients off and those who were still on (slight) sedation. In summary, behavioural responses are unlikely, although they cannot be ruled out.

Second, the PEEP-related effect could also be due to responses of the chemical feedback system. Based on both alveolar gas exchange (and thus on functional residual capacity as a function of PEEP) and the circulatory time (being approximately 6–7 s at normal conditions), any change in alveolar gas composition will influence the response of peripheral (mainly oxygen) and central (mainly carbon dioxide) chemoreceptors [3, 8, 24]. In the absence of lung overdistension the PEEP-related increase in end-expiratory lung volume would ameliorate gas exchange (if there were any effect on gas exchange at all), which then might result in a compensatory downregulation in ventilation. Effects will be stronger in patients with normal than in those with decreased lung compliance. Although such a possibility would fit well the findings of our study, the slight but significant decrease in expired volume of carbon dioxide with increasing levels of PEEP (resulting from the PEEP-related decrease in VE) does not support this hypothesis. In addition, if predominantly chemical feedback responses were at work, then different responses would be expected for the different periods of time for which the PEEP levels were applied (i.e. slight responses with durations of 10 s but stronger responses with durations of 60 s). Finally, the occurrence of the effect within the first breath after PEEP adjustment could hardly be accounted for by chemical feedback mechanisms.

A third alternative explanation for the observed PEEP-related effect might relate to persistent inspiratory muscle activity during exhalation [29, 30]. The presence of persistent inspiratory muscle activity would prolong the expiratory time either independent of or in addition to the reflex-related response. The effect of persistent inspiratory muscle activity would manifest as a depressing influence on expiratory flow rate. However, because expiratory flow (i.e. peak flow rate and expiratory volume) was unaffected by the level of PEEP (Table 3), any hypothesis based on persistent inspiratory muscle activity as the source of the observed PEEP-related effect must be rejected.

In conclusion, the PEEP-related short-term effect on breathing pattern found in the present is best explained by neuronal reflex mechanisms (i.e. the Hering–Breuer inflation/deflation reflex).

In the post hoc analysis we found a similar response to changes of PEEP in patients with normal and those with decreased respiratory system compliance. For the latter, however, the effects were stronger. Consistent with our results, Tryfon and coworkers [14] found a less sensitive reflex in patients with supranormal compliance (i.e. patients with chronic obstructive pulmonary disease) as compared with control individuals (i.e. with relatively lower compliance) or patients with interstitial fibrosis (i.e. with decreased compliance). At first glance these findings are surprising, because with higher respiratory system compliance alterations in PEEP should result in stronger changes in lung volume, and so stronger PEEP-related effects on breathing pattern are anticipated in this setting. The findings of our and Tryfon's study refute this, which might be related to the expiratory muscle recruitment that has been found in awake (but not sleeping) subjects in response to increased end-expiratory lung volume [29, 31, 32]. Expiratory muscle recruitment due to a PEEP-related increase in end-expiratory lung volume would be expected predominantly at normal rather than at decreased respiratory system compliance. The finding of attenuated PEEP-related effects in patients with normal compliance fit well with this hypothesis. Even if some of our patients were under slight sedation (but easy to awake) during the investigation, the hypothesis might hold true because we did not find any difference in PEEP-related effect between wakeful patients and those under slight sedation.

An alternative hypothesis for our unexpected results centres on the potential occurrence of intrinsic PEEP during the investigation. If this is the case, then intrinsic PEEP would have occurred predominantly in patients with normal rather than in those with decreased respiratory system compliance. Consequently, the increase in external PEEP would have reduced the work of breathing, and thus would have attenuated changes in the PEEP-related effect on breathing pattern predominantly in patients with normal respiratory system compliance (as found in our and Tryfon's study). However, there was no evidence for the occurrence of intrinsic PEEP, at least during the preceding period of controlled mechanical ventilation, in which VE was at a similar level as during the observational period of the study. In addition, careful examination of the expiratory flow pattern during the investigation did not suggest expiratory flow limitation as an indirect sign of the occurrence of intrinsic PEEP.

Conclusion

In tracheally intubated, spontaneously breathing adult patients, the level of PEEP significantly influences the resting short-term breathing pattern by selectively affecting expiratory time. The mechanism is probably based on the Hering–Breuer inflation/deflation reflex. Further studies are needed to address counteracting behavioural and/or slow responses of chemical respiratory control, and therefore to elucidate the clinical importance of our findings.

Key messages

  • In spontaneous breathing patients upwards steps in PEEP significantly decrease respiratory rate and minute ventilation whereas downward steps have just opposite effects

  • The PEEP-related effects are exclusively due to alteration of the expiratory time

  • Effects become fully established within the first two breaths after PEEP adjustment and went on for minimally one minute

  • Findings of this study and theoretical considerations strongly suggest a reflex related response by the Hering-Breuer inflation/deflation reflex

Abbreviations

ETT:

endotracheal tube

FiO:

fractional inspired oxygen

PEEP:

positive end-expiratory pressure

PSV:

pressure support ventilation

V:

minute ventilation

V:

tidal volume.

References

  1. Issa FG, Porostocky S: Effect of sleep on changes in breathing pattern accompanying sigh breaths. Respir Physiol 1993, 93: 175-187. 10.1016/0034-5687(93)90004-T

    Article  CAS  PubMed  Google Scholar 

  2. Rabbette PS, Costeloe KL, Stocks J: Persistence of the Hering-Breuer reflex beyond the neonatal period. J Appl Physiol 1991, 71: 474-480.

    CAS  PubMed  Google Scholar 

  3. Coleridge HM, Coleridge JCG: Reflexes evoked from tracheobronchial tree and lungs. In Handbook of Physiology, Section 3: the Respiratory System, Control of Breathing, Part 1. Volume II. Baltimore, MA: The Williams & Wilkins Company; 1986:395-430.

    Google Scholar 

  4. Martin RJ, Nearman HS, Katona PG, Klaus MH: The effect of a low continuous positive airway pressure on the reflex control of respiration in the preterm infant. J Pediatr 1977, 90: 976-981.

    Article  CAS  PubMed  Google Scholar 

  5. Griffin F, Greenough A, Naik S: The Hering-Breuer reflex in ventilated children. Respir Med 1996, 90: 463-466. 10.1016/S0954-6111(96)90172-9

    Article  Google Scholar 

  6. Greenough A, Pool J: Hering-Breuer reflex in young asthmatic children. Pediatr Pulmonol 1991, 11: 345-349.

    Article  CAS  PubMed  Google Scholar 

  7. Stanley NN, Altose MD, Cherniack NS, Fishman AP: Changes in the strength of lung inflation reflex during prolonged inflation. J Appl Physiol 1975, 38: 474-480.

    CAS  PubMed  Google Scholar 

  8. Widdicombe JG: Reflex control of breathing. In MTP International Review of Science, Physiology Series One, Respiratory Physiology. Volume 3. Edited by: Widdicombe JG. London: Butterworth; 1974:273-301.

    Google Scholar 

  9. Rabatte PS, Stocks J: Influence of volume dependency and timing of airway occlusions on the Hering-Breuer reflex in infants. J Appl Physiol 1998, 85: 2033-2039.

    Google Scholar 

  10. Hering E, Breuer J: Self-control of breathing by the vagus nerve [in German]. Sitzber Deut Akad Wiss Wien 1868, 57: 672-690.

    Google Scholar 

  11. Hamilton RD, Winning AJ, Horner RL, Guz A: The effect of lung inflation on breathing in man during wakefulness and sleep. Respir Physiol 1988, 73: 145-154. 10.1016/0034-5687(88)90062-X

    Article  CAS  PubMed  Google Scholar 

  12. Graves C, Glass L, Laporta D, Meloche R, Grassino A: Respiratory phase locking during mechanical ventilation in anesthetized human subjects. Am J Physiol 1986, 250: R902-R909.

    CAS  PubMed  Google Scholar 

  13. Simon PM, Zurob AS, Wies WM, Leiter JC, Hubmayr RD: Entrainment of respiration in humans by periodic lung inflations. Effect of state and CO 2 . Am J Respir Crit Care Med 1999, 160: 950-960.

    Article  CAS  PubMed  Google Scholar 

  14. Tryfon S, Kontakiotis T, Mavrofridis E, Patakas D: Hering-Breuer reflex in normal adults and in patients with chronic obstructive pulmonary disease and interstitial fibrosis. Respiration 2001, 68: 140-144. 10.1159/000050483

    Article  CAS  PubMed  Google Scholar 

  15. Ramsay MAE, Savege TM, Simpson BRJ, Goodwin R: Controlled sedation with alphaxalone-alphadalone. Br Med J 1974, 2: 656-659.

    Article  PubMed Central  CAS  PubMed  Google Scholar 

  16. Fabry B, Guttmann J, Eberhard L, Wolff G: Automatic compensation of endotracheal tube resistance in spontaneously breathing patients. Technol Health Care 1994, 1: 281-291.

    CAS  PubMed  Google Scholar 

  17. Elsasser S, Guttmann J, Stocker R, Mols G, Haberthür C: Accuracy of automatic tube compensation in new-generation mechanical ventilators. Crit Care Med 2003, 31: 2619-2626. 10.1097/01.CCM.0000094224.78718.2A

    Article  PubMed  Google Scholar 

  18. Haberthür C, Fabry B, Stocker R, Ritz R, Guttmann J: Additional inspiratory work of breathing imposed by tracheostomy tubes and non-ideal ventilator properties in critically ill patients. Intensive Care Med 1999, 25: 514-519. 10.1007/s001340050890

    Article  PubMed  Google Scholar 

  19. Guttmann J, Eberhard L, Fabry B, Bertschmann W, Wolff G: Continuous calculation of intratracheal pressure in tracheally intubated patients. Anesthesiology 1993, 79: 503-513.

    Article  CAS  PubMed  Google Scholar 

  20. Haberthür C, Lichwarck-Aschoff M, Guttmann J, Guttmann J: Continuous monitoring of tracheal pressure including spot-check of endotracheal tube resistance. Technol Health Care 2003, 11: 413-424.

    PubMed  Google Scholar 

  21. Guttmann J, Eberhard L, Fabry B, Zappe D, Bernhard H, Lichtwarck-Aschoff M, Adolph M, Wolff G: Determination of volume-dependent respiratory system mechanics in mechanically ventilated patients using the new SLICE method. Technol Health Care 1994, 2: 175-191.

    CAS  PubMed  Google Scholar 

  22. Eberhard L, Guttmann J, Wolff G, Bertschmann W, Minzer A, Kohl HJ, Zeravik J, Adolph M, Eckart J: Intrinsic PEEP monitored in the ventilated ARDS patient with a mathematical method. J Appl Physiol 1992, 73: 479-485.

    CAS  PubMed  Google Scholar 

  23. Brunner JX, Wolff G: Pulmonary Function Indices in Critical Care Patients. Heidelberg: Springer; 1988.

    Book  Google Scholar 

  24. Nilsestuen JO, Coon RL, Woods M, Kampine JP: Location of lung receptors mediating the breathing frequency response to pulmonary CO 2 . Respir Physiol 1981, 45: 343-355. 10.1016/0034-5687(81)90017-7

    Article  CAS  PubMed  Google Scholar 

  25. Iotti GA: Closed-loop control mechanical ventilation. Respir Care N Am 2001, 7: 341-522. 10.1016/S1078-5337(05)70040-X

    Article  Google Scholar 

  26. Cunningham DJC, Robbins PA, Wolff CB: Integration of respiratory responses to changes in alveolar partial pressures of CO 2 and O 2 and in arterial pH. In Handbook of Physiology, Section 3: the Respiratory System, Control of Breathing, Part 2. Volume II. Baltimore, MA: The Williams & Wilkins Company; 1986:475-528.

    Google Scholar 

  27. BuSha BF, Judd BG, Manning HL, Simon PM, Searle BC, Daubenspeck JA, Leiter JC: Identification of respiratory vagal feedback in awake normal subjects using pseudorandom unloading. J Appl Physiol 2001, 90: 2330-2340.

    CAS  PubMed  Google Scholar 

  28. Grunstein MM, Wyszogrodinski I, Milic-Emili J: Regulation of frequency and depth of breathing during expiratory threshold loading in cats. J Appl Physiol 1975, 38: 869-874.

    CAS  PubMed  Google Scholar 

  29. Yan S, Kayser B, Tobiasz M, Sliwinski P: Comparison of static and dynamic intrinsic positive end-expiratory pressure using the Campbell diagram. Am J Respir Crit Care Med 1996, 154: 938-944.

    Article  CAS  PubMed  Google Scholar 

  30. Zakynthinos SG, Vassilakopoulos T, Zakynthinos E, Mavrommatis A, Roussos C: Contribution of expiratory muscle pressure to dynamic intrinsic positive end-expiratory pressure: validation using the Campbell diagram. Am Respir Crit Care Med 2000, 162: 1633-1644.

    Article  CAS  Google Scholar 

  31. Road JD, Leevers AM: Inspiratory and expiratory muscle function during continuous positive airway pressure in dogs. J Appl Physiol 1990, 68: 1092-1100.

    CAS  PubMed  Google Scholar 

  32. Wakai Y, Welsh MM, Leevers AM, Road JD: Expiratory muscle activity in the awake and sleeping human during lung inflation and hypercapnia. J Appl Physiol 1992, 72: 881-887.

    CAS  PubMed  Google Scholar 

Download references

Acknowledgements

The study was supported by a grant from the 'Stiftung Krokus', Basel, Switzerland.

Author information

Authors and Affiliations

Authors

Corresponding author

Correspondence to Christoph Haberthür.

Additional information

Competing interests

The author(s) declare that they have no competing interests.

Authors' contributions

CH designed the study, carried out the measurements, performed the statistical analysis, and drafted the manuscript. JG conceived the study, and participated in its design and helped to draft the manuscript. All authors read and approved the final manuscript.

Authors’ original submitted files for images

Rights and permissions

Reprints and permissions

About this article

Cite this article

Haberthür, C., Guttmann, J. Short-term effects of positive end-expiratory pressure on breathing pattern: an interventional study in adult intensive care patients. Crit Care 9, R407 (2005). https://doi.org/10.1186/cc3735

Download citation

  • Received:

  • Revised:

  • Accepted:

  • Published:

  • DOI: https://doi.org/10.1186/cc3735

Keywords