- Research
- Open access
- Published:
Comparison of two percutaneous tracheostomy techniques, guide wire dilating forceps and Ciaglia Blue Rhino: a sequential cohort study
Critical Care volume 8, Article number: R299 (2004)
Abstract
Introduction
To evaluate and compare the peri-operative and postoperative complications of the two most frequently used percutaneous tracheostomy techniques, namely guide wire dilating forceps (GWDF) and Ciaglia Blue Rhino (CBR).
Methods
A sequential cohort study with comparison of short-term and long-term peri-operative and postoperative complications was performed in the intensive care unit of the University Medical Centre in Nijmegen, The Netherlands. In the period 1997–2000, 171 patients underwent a tracheostomy with the GWDF technique and, in the period 2000–2003, a further 171 patients with the CBR technique. All complications were prospectively registered on a standard form.
Results
There was no significant difference in major complications, either peri-operative or postoperative. We found a significant difference in minor peri-operative complications (P < 0.01) and minor late complications (P < 0.05).
Conclusion
Despite a difference in minor complications between GWDF and CBR, both techniques seem equally reliable.
Introduction
Tracheostomy is usually performed in patients who need prolonged mechanical ventilation, frequent suctioning of bronchopulmonary toilet or have obstruction of the upper airway. The percutaneous tracheostomy is a minimally invasive, effective and reliable procedure and has become the alternative to surgical tracheostomy [1]. Almost all percutaneous procedures in The Netherlands are performed with one of the three following techniques: guide wire dilating forceps (GWDF) tracheostomy, Ciaglia Blue Rhino (CBR) tracheostomy, and sequential dilation tracheostomy (classic Ciaglia) [2]. We have extensive experience with the first two techniques [3, 4]. This study is a sequel to our previous reports. Several studies have compared different percutaneous techniques [5–12], but because CBR is relatively new, a comparison with GWDF has been made only twice in two small prospective cohorts [5, 12]. The strength of the present study is the large group of patients, so the incidence of relevant complications is more meaningful.
The aim of this study was to compare GWDF and CBR. The study not only focuses on the immediate peri-operative complications but also describes the long-term sequelae of both techniques.
Methods
This is a retrospective analysis of all patients who underwent percutaneous tracheostomy in the University Medical Centre Nijmegen between March 1997 and April 2003. We compared the two historic data sets that we have published previously [3, 4], but we specifically focused on the precise definition of early complications and long-term sequelae. Between March 1997 and February 2000 we performed percutaneous tracheostomy on 171 patients, using the GWDF technique. Between March 2000 and April 2003 we performed percutaneous tracheostomy on a further 171 patients, using the CBR technique. Indications, contra-indications and technique for percutaneous tracheostomy are standardised [3, 4]. Patients or family gave informed consent before the procedure. Ethical approval from the institution's medical ethical committee was not obtained because the standard of care was provided and no other experimental treatments were introduced. Published data cannot be reduced to a single recognisable patient. All data were recorded prospectively on pre-designed forms. 'Procedure time' was defined as the time from incision to successful placement of the cannula. A 'peri-operative complication' was defined as a complication related to the procedure and occurring during or within 24 hours of the procedure. Postoperative complications were divided into 'complications while cannulated' and 'late complications'. A 'complication while cannulated' was defined as a complication occurring in the period between 24 hours after the procedure until removal of the cannula. A 'late complication' was defined as a complication occurring after removal of the cannula up to a follow-up of 3 years. Complications were divided into minor and major (see Tables 1, 2, 3). Moreover, complications were classified as procedure-specific and procedure-non-specific. Hypotension was defined as a systolic blood pressure of less than 90 mmHg. Hypoxaemia was defined as an arterial oxygen saturation of less than 90%. It was considered minor when lasting less than 5 min, and major when lasting 5 min or longer. Information regarding late complications was obtained by structured interviews with patients who were decannulated successfully. Patients or close relatives were asked about voice changes, dyspnoea, stridor, pain, and cosmetic problems. Patients were also asked to grade specific problems as absent, minor or major.
All data were analysed with Statistical Product and Service Solutions (SPSS) version 11.0. All variables were checked for normal distribution. Data are given as means ± SD or medians. Continuous variables were compared with Student's t-test or the Mann–Whitney test as appropriate. Bonferroni's correction for multiple comparisons was used. Categorisable variables were compared with the χ2 test. A cut-off level of P < 0.05 was accepted as statistically significant.
Results
Demographic data are shown in Table 4. The procedure was successful in 165 of 171 patients (96.5%) in the GWDF group and in 169 of 171 patients (98.8%) in the CBR group. Most tracheostomies were performed by an intensivist or a fellow (under supervision). More procedures were performed by a fellow in the CBR group than in the GWDF group (51 versus 27, respectively; P < 0.01).
Peri-operative complications
Peri-operative complications are described in Table 1. In total, there were 47 peri-operative complications in 43 patients in the GWDF group, and 84 peri-operative complications in 71 patients in the CBR group (P < 0.05). This difference is explained by a greater number of difficult dilations (P < 0.01) and minor bleedings with the CBR technique. After the introduction of a Crile's forceps for blunt dissection of the pretracheal tissues preceding CBR, the procedure became much easier. In the GWDF group, 13 patients (7.6%) had a major complication, compared with 9 patients (5.3%) in the CBR group. All these major peri-operative complications were procedure-specific. One life-threatening bleeding in the GWDF group led to severe hypoxia at the end of the procedure. After removal of the cannula, large blood clots were suctioned from the trachea. There was no significant difference in the number of patients in whom conversion to a surgical tracheostomy was necessary. In the GWDF group, six patients underwent conversion to a surgical tracheostomy: one patient had a major venous bleeding after dilation of the trachea and the cannula could not be inserted. In another patient, arterial blood was aspirated and the procedure was terminated. In two patients, the trachea was difficult to locate, resulting in hypoxaemia and hypercapnia. In one patient the guide wire was placed correctly but the cannula perforated the posterior tracheal wall and entered the oesophagus. Surgical exploration confirmed rupture of the oesophagus, and the tracheo-oesophageal wall was immediately repaired. The post-operative course was uneventful. In the last patient the distance between skin and trachea was too large for the insertion of a cannula. In the CBR group two patients underwent surgical tracheostomy: in one patient the trachea was difficult to locate, and the cannula was placed pretracheally as a result of guide wire kinking. Another patient developed major bleeding and tension pneumothorax several hours after the procedure. After immediate drainage with a chest tube, surgical exploration showed that the tracheostomy tube had perforated the cricothyroid membrane. No deaths were seen after either procedure.
Complications while cannulated
In total, 164 GWDF and 169 CBR patients were analysed for complications while cannulated (Table 2). Four major complications (2.4%) occurred in the GWDF group, and seven major complications (4.1%) in the CBR group. One patient in the GWDF group had an obstruction of the cannula by a mucous plug, leading to a cardiorespiratory arrest. Another patient sustained a cardiorespiratory arrest shortly after decannulation, possibly due to aspiration. Both patients were resuscitated successfully. Three patients in the CBR group had an obstruction of the cannula: one of them died on his first day on the ward, possibly owing to an obstructive blood clot in the cannula. The second patient had a mucous plug causing severe hypoxaemia. He received a minitracheotomy through the old tracheostomy opening. The third patient with an obstructed cannula was found in bed on the ward, having a respiratory arrest. The inner cannula, which was obstructed by a blood clot, was removed. The patient recovered uneventfully.
Late complications
Of 164 patients in the GWDF group, 53 (32.3%) died with the cannula in place or within 1 week after decannulation, and five patients were lost to follow-up. One hundred and seven GWDF patients (62.6%) were decannulated successfully and analysed for late complications (Table 3). Of 169 CBR patients, 60 (35.5%) died with the cannula in place or within 1 week of decannulation, six patients were lost to follow-up, and three patients had the cannula still in situ. Finally, 100 CBR patients (58.5%) were analysed for late complications. There was no significant difference between both groups with regard to total late complications. All patients with voice problems were given the opportunity to consult an ENT specialist. None of these had an objective laryngeal abnormality explaining their voice problems. Patients with cosmetic problems relating to the tracheostomy scar were offered specialist consultation. Six GWDF patients underwent scar revision. Three patients developed a severe stridor after decannulation. In the GWDF group, an 83-year-old woman had tracheal stenosis and was treated with an endotracheal stent, and an 80-year-old woman was treated with laser for a granuloma just above the tracheostomy opening. In the CBR group, an 18-year-old man suffered from severe tracheal stenosis. He had a tracheal stent placed initially, but because of recurrence of the stenosis, a tracheal resection was necessary. The patient recovered uneventfully.
Discussion
In this study we have compared two different techniques of percutaneous tracheostomy, GWDF and CBR. Both techniques are frequently used in The Netherlands and are replacing the surgical technique [2]. This study showed no significant differences in clinically relevant complications between the two techniques. This is in agreement with two other studies comparing these techniques [5, 12]. Although the total number of complications in the two groups in the study of Ambesh and colleagues was not significantly different, the authors noticed an increased rate of minor peri-operative bleeding in the GDWF group [5]. This was balanced by an increase in the number of patients with one or more tracheal ring fractures in the CBR group (30%). The increase in major peri-operative bleeding with the GDWF technique might be explained by the poorly controllable dilation with the forceps [9]. Although the study of Añón and colleagues did not find any significant differences, in three of 26 patients in the GWDF group there was an inability to insert the cannula [12].
Several other studies comparing sequential dilation (classic Ciaglia) and CBR [6, 8], and comparing sequential dilation and GWDF [7, 9–11], have been described in the literature. Van Heurn and colleagues concluded that sequential dilation and GWDF are both reliable but that sequential dilation has fewer early complications than GWDF [7]. Nates and colleagues also preferred sequential dilation to the GWDF technique, because of fewer surgical complications, less peri-operative and postoperative bleeding, and easier use [9]. Añón and colleagues found a comparable complication rate, but the procedural time of the GWDF method was significantly shorter [10]. Unfortunately, comparing these studies is difficult because complications were not defined uniformly.
In our study, a major complication while cannulated was obstruction of the cannula, which occurred in four patients. These figures correspond to the prevalence of cannula obstruction in the literature (0.3–3.5%) [13–15]. Strict adherence to nursing protocols and a low threshold for cleaning the inner cannula should be the standard of care in the intensive care unit. An outreach team from the intensive care unit should visit patients, discharged to the general ward with a cannula in place, on a daily basis.
There are only few data available concerning late complications of percutaneous tracheostomy. Unfortunately, many confounders might be present, such as the disease process itself, the duration of endotracheal intubation, and other treatments in the intensive care unit (such as sedation or physical therapy). Moreover, both patients and caregivers often interpret late complications subjectively. The total number of late complications in our study was not significantly different between the two groups. Subjective voice changes and hoarseness were more frequent in the CBR group (P < 0.01). An explanation might be the longer mean endotracheal intubation time, because this is possibly the most important cause of voice problems. With sequential dilation tracheostomy, the incidence of voice problems ranges between 0% and 21% [16–22]. More patients in the GWDF group complained of cosmetic problems. Only a few studies have mentioned cosmetic complaints, but differences of opinion between patient and caregiver are frequent [23]. In each group in our study, one patient developed a critical symptomatic tracheal stenosis. More patients might have had an asymptomatic tracheal stenosis, but because no additional diagnostic tests such as computed tomography or magnetic resonance imaging scans were performed, the actual incidence is unknown. Several studies have incriminated the GWDF technique as a cause of tracheal stenosis, but no studies with the CBR have been described. The incidence varied from 0% to 63% [18, 23–27]. Most of these tracheal stenoses were asymptomatic.
Several factors might decrease the strength of our conclusions. First, the study used historical data sets with a sequential design; a time bias is therefore possible. As experience with percutaneous tracheostomy increases, the number of complications will decrease, even if another technique is used, although in our study this might well have been balanced by the fact that over time more fellows performed the procedure. Second, scoring of the peri-operative complications by different physicians might be variable because of different interpretations. Despite these shortcomings, we conclude from our study that, although the CBR technique has more minor peri-operative complications, the two techniques are comparable. More prospective, randomised studies are required to compare these different tracheostomy techniques adequately. We are currently conducting a prospective, randomised study in which we compare GWDF and CBR tracheostomies; we are specifically looking for the occurrence of precisely defined early and late complications. The occurrence of tracheal stenosis will be analysed using the forced oscillation technique and magnetic resonance imaging.
Key messages
-
GWDF and CBR tracheostomy seem equally reliable.
-
Major peri-operative complications occur in 5.3–7.6% of patients.
-
Late complications are rare
Abbreviations
- CBR:
-
CBR = Ciaglia Blue Rhino
- GWDF:
-
GWDF = guide wire dilating forceps.
References
Friedman Y, Fildes J, Mizock B, Samuel J, Patel S, Appavu S, Roberts R: Comparison of percutaneous and surgical tracheostomies. Chest 1996, 110: 480-485.
Fikkers BG, Fransen GA, van der Hoeven JG, Briede IS, van den Hoogen FJ: Tracheostomy for long-term ventilated patients: a postal survey of ICU practice in the Netherlands. Intensive Care Med 2003, 29: 1390-1393. 10.1007/s00134-003-1824-x
Fikkers BG, van Heerbeek N, Krabbe PF, Marres HA, van den Hoogen FJ: Percutaneous tracheostomy with the guide wire dilating forceps technique: presentation of 171 consecutive patients. Head Neck 2002, 24: 625-631. 10.1002/hed.10113
Fikkers BG, Briede IS, Verwiel JM, van den Hoogen FJ: Percutaneous tracheostomy with the Blue Rhino technique: presentation of 100 consecutive patients. Anaesthesia 2002, 57: 1094-1097. 10.1046/j.1365-2044.2002.02834.x
Ambesh SP, Pandey CK, Srivastava S, Agarwal A, Singh DK: Percutaneous tracheostomy with single dilatation technique: a prospective, randomized comparison of Ciaglia Blue Rhino versus Griggs' guidewire dilating forceps. Anesth Analg 2002, 95: 1739-1745. 10.1097/00000539-200212000-00050
Johnson JL, Cheatham ML, Sagraves SG, Block EF, Nelson LD: Percutaneous dilational tracheostomy: a comparison of single- versus multiple-dilator techniques. Crit Care Med 2001, 29: 1251-1254. 10.1097/00003246-200106000-00036
van Heurn LW, Mastboom WB, Scheeren CI, Brink PR, Ramsay G: Comparative clinical trial of progressive dilatational and forceps dilatational tracheostomy. Intensive Care Med 2001, 27: 292-295. 10.1007/s001340000743
Byhahn C, Wilke HJ, Halbig S, Lischke V, Westphal K: Percutaneous tracheostomy: Ciaglia Blue Rhino versus the basic Ciaglia technique of percutaneous dilational tracheostomy. Anesth Analg 2000, 91: 882-886. 10.1097/00000539-200010000-00021
Nates JL, Cooper DJ, Myles PS, Scheinkestel CD, Tuxen DV: Percutaneous tracheostomy in critically ill patients: a prospective, randomized comparison of two techniques. Crit Care Med 2000, 28: 3734-3739. 10.1097/00003246-200011000-00034
Añón JM, Gómez V, Escuela MP, De Paz V, Solana LF, De La Casa RM, Pérez JC, Zeballos E, Navarro L: Percutaneous tracheostomy: comparison of Ciaglia and Griggs techniques. Crit Care 2000, 4: 124-128. 10.1186/cc667
Ambesh SP, Kaushik S: Percutaneous dilational tracheostomy: the Ciaglia method versus the Portex [correction of Rapitrach] method. Anesth Analg 1998, 87: 556-561. 10.1097/00000539-199809000-00010
Anon JM, Escuela MP, Gomez V, Moreno A, Lopez J, Diaz R, Montejo JC, Sirgo G, Hernandez G, Martinez R: Percutaneous tracheostomy: Ciaglia Blue Rhino versus Griggs' guide wire dilating forceps. A prospective randomized trial. Acta Anaesthesiol Scand 2004, 48: 451-456.
Trottier SJ, Ritter S, Lakshmanan R, Sakabu SA, Troop BR: Percutaneous tracheostomy tube obstruction: warning. Chest 2002, 122: 1377-1381. 10.1378/chest.122.4.1377
Powell DM, Price PD, Forrest LA: Review of percutaneous tracheostomy. Laryngoscope 1998, 108: 170-177. 10.1097/00005537-199802000-00004
Dulguerov P, Gysin C, Perneger TV, Chevrolet JC: Percutaneous or surgical tracheostomy: a meta-analysis. Crit Care Med 1999, 27: 1617-1625. 10.1097/00003246-199908000-00041
Ciaglia P, Graniero KD: Percutaneous dilatational tracheostomy. Results and long-term follow-up. Chest 1992, 101: 464-467.
Walz MK, Peitgen K, Thurauf N, Trost HA, Wolfhard U, Sander A, Ahmadi C, Eigler FW: Percutaneous dilatational tracheostomy – early results and long-term outcome of 326 critically ill patients. Intensive Care Med 1998, 24: 685-690. 10.1007/s001340050645
van Heurn LW, Goei R, de Ploeg I, Ramsay G, Brink PR: Late complications of percutaneous dilatational tracheotomy. Chest 1996, 110: 1572-1576.
Marelli D, Paul A, Manolidis S, Walsh G, Odim JN, Burdon TA, Shennib H, Vestweber KH, Fleiszer DM, Mulder DS: Endoscopic guided percutaneous tracheostomy: early results of a consecutive trial. J Trauma 1990, 30: 433-435.
Hill BB, Zweng TN, Maley RH, Charash WE, Toursarkissian B, Kearney PA: Percutaneous dilational tracheostomy: report of 356 cases. J Trauma 1996, 41: 238-243.
Fischler MP, Kuhn M, Cantieni R, Frutiger A: Late outcome of percutaneous dilatational tracheostomy in intensive care patients. Intensive Care Med 1995, 21: 475-481.
Law RC, Carney AS, Manara AR: Long-term outcome after percutaneous dilational tracheostomy. Endoscopic and spirometry findings. Anaesthesia 1997, 52: 51-56. 10.1111/j.1365-2044.1997.013-az013.x
Sviri S, Samie R, Roberts BL, Van Heerden PV: Long-term outcomes following percutaneous tracheostomy using the Griggs technique. Anaesth Int Care 2003, 31: 401-407.
Norwood S, Vallina VL, Short K, Saigusa M, Fernandez LG, McLarty JW: Incidence of tracheal stenosis and other late complications after percutaneous tracheostomy. Ann Surg 2000, 232: 233-241. 10.1097/00000658-200008000-00014
Dollner R, Verch M, Schweiger P, Deluigi C, Graf B, Wallner F: Laryngotracheoscopic findings in long-term follow-up after Griggs tracheostomy. Chest 2002, 122: 206-212. 10.1378/chest.122.1.206
Steele AP, Evans HW, Afaq MA, Robson JM, Dourado J, Tayar R, Stockwell MA: Long-term follow-up of Griggs percutaneous tracheostomy with spiral CT and questionnaire. Chest 2000, 117: 1430-1433. 10.1378/chest.117.5.1430
Leonard RC, Lewis RH, Singh B, Van Heerden PV: Late outcome from percutaneous tracheostomy using the Portex kit. Chest 1999, 115: 1070-1075. 10.1378/chest.115.4.1070
Author information
Authors and Affiliations
Corresponding author
Additional information
Competing interests
None declared.
Rights and permissions
About this article
Cite this article
Fikkers, B.G., Staatsen, M., Lardenoije, S.G. et al. Comparison of two percutaneous tracheostomy techniques, guide wire dilating forceps and Ciaglia Blue Rhino: a sequential cohort study. Crit Care 8, R299 (2004). https://doi.org/10.1186/cc2907
Received:
Accepted:
Published:
DOI: https://doi.org/10.1186/cc2907