Impact of cardiac output and alveolar ventilation in estimating ventilation/perfusion mismatch in ARDS using electrical impedance tomography

Abstract

Introduction

Electrical impedance tomography (EIT) can be used to assess ventilation/perfusion (V/Q) mismatch within the lungs. Several methods have been proposed, some of them neglecting the absolute value of alveolar ventilation (VA) and cardiac output (QC). Whether this omission results in acceptable bias is unknown.

Methods

Pixel-level V/Q maps of 25 ARDS patients were computed once considering (absolute V/Q map) and once neglecting (relative V/Q map) the value of QC and VA. Previously published indices of V/Q mismatch were computed using absolute V/Q maps and relative V/Q maps. Indices computed with relative V/Q maps were compared to their counterparts computed using absolute V/Q maps.

Results

Among 21 patients with ratio of alveolar ventilation to cardiac output (VA/QC) > 1, relative shunt fraction was significantly higher than absolute shunt fraction [37% (24–66) vs 19% (11–46), respectively, p < 0.001], while relative dead space fraction was significantly lower than absolute dead space fraction [40% (22–49) vs 58% (46–84), respectively, p < 0.001]. Relative wasted ventilation was significantly lower than the absolute wasted ventilation [16% (11–27) vs 29% (19–35), respectively, p < 0.001], while relative wasted perfusion was significantly higher than absolute wasted perfusion [18% (11–23) vs 11% (7–19), respectively, p < 0.001]. The opposite findings were retrieved in the four patients with VA/QC < 1.

Conclusion

Neglecting cardiac output and alveolar ventilation when assessing V/Q mismatch indices using EIT in ARDS patients results in significant bias, whose direction depends on the VA/QC ratio value.

Introduction

Acute respiratory distress syndrome (ARDS) is associated with lung ventilation and perfusion mismatch, whose severity is associated with poor prognosis [1]. Electrical impedance tomography (EIT) can be used to assess distribution of both ventilation and perfusion within the lungs [2]. Several methods for calculating pixel-level V/Q ratios have been reported [2, 3]. We proposed a method based on the calculation of the ventilation and perfusion received by each pixel, allowing a pixel by pixel calculation of the "absolute" pixel-level V/Q ratio [4]. More recently, Pavlovski et al. reported a method also based on a pixel-by-pixel analysis, but without considering actual cardiac output (QC) and alveolar ventilation (VA) in the calculation of the pixel-level V/Q ratio [5], which were instead calculated from the relative participation of each pixel in ventilation and perfusion, implicitly considering that cardiac output and alveolar ventilation are equal. Although this method has the advantage of allowing estimation of pixel-level V/Q ratios with EIT data alone, whether it results in acceptable bias is unknown.

Our objective was to evaluate the impact of neglecting cardiac output and alveolar ventilation when assessing ventilation/perfusion mismatch indices using EIT.

Methods

This is an ancillary and retrospective analysis of data from patients included in a prospective single-center study (ethic agreement: CPP-66/17) investigating the impact of PEEP variations and prone position on ventilation and perfusion distributions [4]. Recordings made with a PEEP level of 12 cmH2O in the supine position were selected for this study.

Patients were intubated, sedated, paralyzed, ventilated in volume control mode with a tidal volume between 6 and 8 mL/kg of predicted body weight (PBW). Ventilation and perfusion distributions were studied using the Enlight 1800 (TIMPEL SA, São Paulo, Brazil). Cardiac output (QC) was measured using the Vigileo device (Edwards Lifesciences, Irvine, CA), in hemodynamically stabilized patients. Alveolar ventilation (VA) was calculated as 70% of the expiratory minute ventilation measured on the mechanical ventilator (in order to remove ventilation of anatomical and instrumental dead space). The ventilation map, which divided the thoracic field into 1024 pixels (32 × 32), was generated using a dedicated software (Timpel offline analysis, TIMPEL SA, São Paulo, Brazil). This map indicated, for each pixel, the fraction of the total alveolar ventilation directed to that pixel. The perfusion map was created as previously described [6]. The perfusion map also included 1024 pixels (32 × 32), corresponding one by one to the pixels of the ventilation map, and indicated, for each pixel, the fraction of the total pulmonary perfusion directed to that pixel.

The pixel-level V/Q ratio was then calculated, for each pixel, by two distinct methods: without taking into account (“relative” pixel-level V/Q) and with taking into account (“absolute” pixel-level V/Q) cardiac output and alveolar ventilation:

  • “relative” pixel-level V/Q \({\left(\frac{V}{Q}\right)}_{\mathrm{PIXEL},\mathrm{REL}}\)

    • The percentage of total ventilation captured by pixel i (V%,i) was divided by the percentage of total perfusion captured by the same pixel (Q%,i), as in [5].

      $${\left(\frac{V}{Q}\right)}_{\mathrm{PIXEL},\mathrm{ REL},i}=\frac{{V}_{{\%},i}}{{Q}_{{\%},i}}:$$
  • “absolute” pixel-level V/Q \({\left(\frac{V}{Q}\right)}_{\mathrm{PIXEL},\mathrm{ ABS}}\)

    • First, the percentage of total ventilation captured by pixel i (V%,i) was multiplied by the total alveolar ventilation, in order to calculate the absolute alveolar ventilation of this pixel

    • Second, the percentage of total perfusion captured by pixel i (Q%,i) was multiplied by the cardiac output, in order to calculate the absolute perfusion of this pixel

    • Third, the absolute pixel-level V/Q ratio was calculated by dividing the absolute ventilation by the absolute perfusion of pixel i:

      $${\left(\frac{V}{Q}\right)}_{\mathrm{PIXEL},\mathrm{ ABS}, i} =\frac{{V}_{{\%},\mathrm{i}}*{V}_{\mathrm{A}}}{{Q}_{{\%},\mathrm{i}}* {Q}_{\mathrm{c}}}$$

To estimate the potential impact of neglecting the alveolar ventilation and cardiac output on the estimation of ventilation/perfusion mismatch, the following variables were computed both using \({\left(\frac{V}{Q}\right)}_{\mathrm{PIXEL},\mathrm{ ABS}}\) and \({\left(\frac{V}{Q}\right)}_{\mathrm{PIXEL},\mathrm{ REL}}\):

  • Distribution of pixels among five different classes of ventilation/perfusion match: shunt pixels (pixel-level V/Q ratio ≤ 0.1); pixels with low V/Q ratio (pixel-level V/Q ratio 0.1–0.8); pixels with normal V/Q ratio (pixel-level V/Q ratio 0.8–1.25); pixels with high V/Q ratio (V/Q ratio 1.25–10); dead space pixels (pixel-level V/Q ratio ≥ 10). These definitions of V/Q unmatching were defined by analogy to [5].

  • Shunt fraction [4]: the fraction of cardiac output reaching pixels with shunt or low V/Q ratio

  • Dead space fraction [4]: the fraction of alveolar ventilation reaching pixels with dead space or high V/Q ratio

  • Wasted ventilation, as defined by [5]

  • Wasted perfusion, as defined by [5]

Value of each variable computed using \({\left(\frac{V}{Q}\right)}_{\mathrm{PIXEL},\mathrm{ ABS}}\)(absolute shunt fraction, absolute dead space fraction, absolute wasted ventilation and absolute wasted perfusion) was compared to the value computed using \({\left(\frac{V}{Q}\right)}_{\mathrm{PIXEL},\mathrm{ REL}}\) (relative shunt fraction, relative dead space fraction, relative wasted ventilation and relative wasted perfusion) by Wilcoxon paired test. Since the direction of the changes in ventilation/perfusion mismatch indices between the two methods (with and without consideration of the cardiac output and alveolar ventilation) is assumed to be opposite between patients with alveolar ventilation superior to cardiac output and those with alveolar ventilation inferior to cardiac output, we planned to analyze these two groups separately. To test the robustness of our results with respect to the threshold defining a normal ventilation/perfusion ratio, we performed a sensitivity analysis, considering a normal ventilation/perfusion ratio between 0.5 and 2. To determine whether our results were robust to the definition of alveolar ventilation, we repeated the analyses by calculating anatomic dead space using two other formulas [7, 8]. More details about definitions and methods are available in the online supplement. Results are reported as median and interquartile range. Statistical analysis was made with GraphPad Prism version 9.4.1 (GraphPad Software, LLC).

Results

Twenty-five ARDS patients were included in the analysis. Detailed patient characteristics are presented in the online supplement (Additional file 1: Table S1). Twenty-one patients exhibited an alveolar ventilation superior to their cardiac output and four an alveolar ventilation inferior to the cardiac output.

Among the 21 patients with alveolar ventilation superior to cardiac output, the \({\left(\frac{V}{Q}\right)}_{\mathrm{PIXELREL}}\) systematically classified more pixels in low V/Q and less pixels in high V/Q compared to \({\left(\frac{V}{Q}\right)}_{\mathrm{PIXELABS}}\) (Table 1). Thus, relative shunt fraction was significantly higher than absolute shunt fraction [37% (24–66) vs 19% (11–46), respectively, p < 0.001; Bland–Altman comparison: bias 14%, LAL − 6%, UAL 33%], while relative dead space fraction was significantly lower than absolute dead space fraction [40% (22–49) vs 58% (46–84), respectively, p < 0.001; Bland–Altman comparison: bias − 27%, LAL − 79%, UAL 25%] (Fig. 1). Relative wasted ventilation was significantly lower than the absolute wasted ventilation [16% (11–27) vs 29% (19–35), respectively, p < 0.001; Bland–Altman comparison: bias − 8%, LAL − 23%, UAL 6%], while relative wasted perfusion was also significantly higher than absolute wasted perfusion [18% (11–23) vs 11% (7–19), respectively, p < 0.001; Bland–Altman comparison: bias 5%, LAL − 1%, UAL 11%]. Bland–Altman plots are available in the supplement (Additional file 1: Figure S1).

Table 1 Comparison of V/Q mismatch indices between absolute and relative V/Q
Fig. 1
figure 1

Differences in ventilation/perfusion matching indices depending on the accounting of cardiac output and alveolar ventilation. A Patients with VA/QC > 1. From left to right: shunt fraction, dead space fraction, wasted perfusion, wasted ventilation. Blue boxes represent absolute indices. Red boxes represent relative indices. Individual patient data are represented by black dots. * denotes p < 0.0001. B Patients with VA/QC < 1. From left to right: shunt fraction, dead space fraction, wasted perfusion, wasted ventilation. Blue boxes represent absolute indices. Red boxes represent relative indices. Individual patient data are represented by black dots. C V/Q maps of a representative patient. Left: absolute V/Q map. Right: relative V/Q map. Blue pixels correspond to high V/Q and dead space pixels. Red pixels correspond to low V/Q and shunt pixels. Pink pixels correspond to pixels with normal V/Q

The opposite findings were retrieved in the four patients with alveolar ventilation inferior to cardiac output (Table 1 and Fig. 1), but the low sample size precluded any statistical analysis.

Difference between relative and absolute wasted ventilation on the one hand, and relative and absolute wasted perfusion on the other hand, were both correlated with the ratio of alveolar ventilation to cardiac output (Additional file 1: Figure S2 and S3).

Sensitivity analysis with a definition of normal pixel-level V/Q ratio between 0.5 and 2 yielded similar results (Additional file 1: Table S2). Similarly, analysis considering other formulas for calculating alveolar ventilation yielded similar results (Additional file 1: Tables S3 and S4).

Discussion

The main finding of our study is that the relative and absolute pixel-level V/Q calculations resulted in significantly different ventilation/perfusion mismatch assessments.

Assessing ventilation/perfusion mismatch may help customizing mechanical ventilation. However, the ventilator settings themselves, in particular the level of PEEP, may alter the cardiac output [9] in a way that is not predictable [10]. Thus, even in a single patient, comparing ventilation/perfusion mismatch indices across multiple ventilatory strategies, neglecting the actual alveolar ventilation to cardiac output ratio, seems hazardous.

Our study has several limitations:

  1. 1.

    The sample size is modest. Nevertheless, measured cardiac output and minute ventilation were consistent with previous reports [11, 12].

  2. 2.

    Alveolar ventilation was not measured but calculated as a fraction of the minute ventilation. However, repeating the analyses with different formulas for estimating the anatomical dead space [7, 8, 13] yielded consistent results.

  3. 3.

    Patients were assessed at a single PEEP level. However, the level chosen (12 cm H2O) was intermediate in order not to promote extreme values of alveolar ventilation to cardiac output ratios.

  4. 4.

    The thresholds defining pixel-level V/Q mismatch were arbitrary. We chose these thresholds to be consistent with the existing literature. Sensitivity analysis using larger thresholds to define normal V/Q yielded similar results. Finally, wasted ventilation and wasted perfusion are insensitive to the thresholds defining abnormal ventilation/perfusion ratios, allowing this limitation to be overcome.

  5. 5.

    There is a theoretical risk of concluding that the ventilation/perfusion ratio is normal at the pixel level, while ventilation/perfusion abnormalities exist at the alveolar level, with opposite anomalies canceling each other out. However, given the spatial resolution of EIT (depending on the patient, between 300 and 500 pixels represent the lung), this risk seems minimal.

In conclusion, neglecting cardiac output and alveolar ventilation when assessing V/Q mismatch indices using EIT in ARDS patients results in significant bias, whose direction depends on the ratio of alveolar ventilation to cardiac output.

Availability of data and materials

The datasets used and/or analyzed during the current study are available from the corresponding author on reasonable request.

Abbreviations

ARDS:

Acute respiratory distress syndrome

EIT:

Electrical impedance tomography

PEEP:

Positive end-expiratory pressure

Q C :

Cardiac output

V A :

Alveolar ventilation

V/Q :

Ventilation to perfusion ratio

References

  1. Nuckton TJ, Alonso JA, Kallet RH, Daniel BM, Pittet JF, Eisner MD, et al. Pulmonary dead-space fraction as a risk factor for death in the acute respiratory distress syndrome. New Engl J Med. 2009. https://doi.org/10.1056/NEJMoa012835.

    Article  Google Scholar 

  2. Xu M, He H, Long Y. Lung perfusion assessment by bedside electrical impedance tomography in critically ill patients. Front Physiol. 2021. https://doi.org/10.3389/fphys.2021.748724.

    Article  PubMed  PubMed Central  Google Scholar 

  3. Spinelli E, Kircher M, Stender B, Ottaviani I, Basile MC, Marongiu I, et al. Unmatched ventilation and perfusion measured by electrical impedance tomography predicts the outcome of ARDS. Crit Care. 2021;25(1):192.

    Article  PubMed  PubMed Central  Google Scholar 

  4. Perier F, Tuffet S, Maraffi T, Alcala G, Victor M, Haudebourg AF, et al. Effect of positive end-expiratory pressure and proning on ventilation and perfusion in COVID-19 acute respiratory distress syndrome. Am J Respir Crit Care Med. 2020;202(12):1713–7.

    Article  CAS  PubMed  PubMed Central  Google Scholar 

  5. Pavlovsky B, Pesenti A, Spinelli E, Scaramuzzo G, Marongiu I, Tagliabue P, et al. Effects of PEEP on regional ventilation-perfusion mismatch in the acute respiratory distress syndrome. Crit Care. 2022;26(1):211.

    Article  PubMed  PubMed Central  Google Scholar 

  6. He H, Chi Y, Long Y, Yuan S, Frerichs I, Möller K, et al. Influence of overdistension/recruitment induced by high positive end-expiratory pressure on ventilation–perfusion matching assessed by electrical impedance tomography with saline bolus. Crit Care. 2020;24(1):586.

    Article  PubMed  PubMed Central  Google Scholar 

  7. Nielsen L. Assessing patients’ respiratory problems. AJN Am J Nurs. 1980;80(12):2192.

    Google Scholar 

  8. Suwa K, Bendixen HH. Change in PaCO2 with mechanical dead space during artificial ventilation. J Appl Physiol. 1968;24(4):556–62.

    Article  CAS  PubMed  Google Scholar 

  9. Suter PM, Fairley HB, Isenberg MD. Optimum end-expiratory airway pressure in patients with acute pulmonary failure. N Engl J Med. 1975;292(6):284–9.

    Article  CAS  PubMed  Google Scholar 

  10. De Santis Santiago R, Teggia Droghi M, Fumagalli J, Marrazzo F, Florio G, Grassi LG, et al. High pleural pressure prevents alveolar overdistension and hemodynamic collapse in ARDS with class III obesity. Am J Respir Crit Care Med. 2020. https://doi.org/10.1164/rccm.201909-1687OC.

    Article  Google Scholar 

  11. Caravita S, Baratto C, Di Marco F, Calabrese A, Balestrieri G, Russo F, et al. Haemodynamic characteristics of COVID-19 patients with acute respiratory distress syndrome requiring mechanical ventilation. An invasive assessment using right heart catheterization. Eur J Heart Fail. 2020;22(12):2228–37.

    Article  CAS  PubMed  Google Scholar 

  12. Bellani G, Laffey JG, Pham T, Fan E, Brochard L, Esteban A, et al. Epidemiology, patterns of care, and mortality for patients with acute respiratory distress syndrome in intensive care units in 50 countries. JAMA. 2016;315(8):788–800.

    Article  CAS  PubMed  Google Scholar 

  13. Brewer LM, Orr JA, Pace NL. Anatomic dead space cannot be predicted by body weight. Respir Care. 2008;53(7):885–91.

    PubMed  Google Scholar 

Download references

Funding

None.

Author information

Authors and Affiliations

Authors

Contributions

ST was involved in study design, data collection, analysis and interpretation, and script writing. TM was involved in study design, data collection, analysis and interpretation. ML was involved in analysis and interpretation. FP was involved in data collection and data analysis. AFH was involved in data collection and script writing. MAB was involved in data collection, analysis, interpretation and script writing. GA was involved in data analysis. AMD was involved in study design, data analysis and interpretation, and script writing. GC was involved in study design, data analysis and interpretation, and script writing. All authors read and approved the final manuscript.

Corresponding author

Correspondence to Samuel Tuffet.

Ethics declarations

Ethics approval and consent to participate

This is an ancillary report of a prospective monocentric observational study on EIT in patients with ARDS (CPP-66/17). The collection and analysis of data provided by electrical impedance tomography had received the approval of the CPP Ile de France VI, on November 8, 2017. Written informed consent was waived due to the observational nature of the study.

Consent for publication

NA.

Competing interests

The authors declare that they have no competing interests.

Additional information

Publisher's Note

Springer Nature remains neutral with regard to jurisdictional claims in published maps and institutional affiliations.

Supplementary Information

Additional file 1

Supplementary appendix: detailed methods and supplemental results.

Rights and permissions

Open Access This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.

Reprints and permissions

About this article

Check for updates. Verify currency and authenticity via CrossMark

Cite this article

Tuffet, S., Maraffi, T., Lacheny, M. et al. Impact of cardiac output and alveolar ventilation in estimating ventilation/perfusion mismatch in ARDS using electrical impedance tomography. Crit Care 27, 176 (2023). https://doi.org/10.1186/s13054-023-04467-w

Download citation

  • Received:

  • Accepted:

  • Published:

  • DOI: https://doi.org/10.1186/s13054-023-04467-w

Keywords