Performance of the cuff leak test in adults in predicting post-extubation airway complications: a systematic review and meta-analysis
Critical Care volume 24, Article number: 640 (2020)
Abstract
Background
Clinical practice guidelines recommend performing a cuff leak test in mechanically ventilated adults who meet extubation criteria to screen those at high risk for post-extubation stridor. Previous systematic reviews demonstrated excellent specificity of the cuff leak test but disagreed with respect to sensitivity. We conducted a systematic review and meta-analysis to assess the diagnostic accuracy of the cuff leak test for predicting post-extubation airway complications in intubated adult patients in critical care settings.
Methods
We searched Medline, EMBASE, Scopus, ISI Web of Science, the Cochrane Library for eligible studies from inception to March 16, 2020, without language restrictions. We included studies that examined the diagnostic accuracy of cuff leak test if post-extubation airway obstruction after extubation or reintubation was explicitly reported as the reference standard. Two authors in duplicate and independently assessed the risk of bias using the Quality Assessment for Diagnostic Accuracy Studies-2 tool. We pooled sensitivities and specificities using generalized linear mixed model approach to bivariate random-effects meta-analysis. Our primary outcomes were post-extubation airway obstruction and reintubation.
Results
We included 28 studies involving 4493 extubations. Three studies were at low risk for all QUADAS-2 risk of bias domains. The pooled sensitivity and specificity of cuff leak test for post-extubation airway obstruction were 0.62 (95% CI 0.49–0.73; I2 = 81.6%) and 0.87 (95% CI 0.82–0.90; I2 = 97.8%), respectively. The pooled sensitivity and specificity of the cuff leak test for reintubation were 0.66 (95% CI 0.46–0.81; I2 = 48.9%) and 0.88 (95% CI 0.83–0.92; I2 = 87.4%), respectively. Subgroup analyses suggested that the type of cuff leak test and length of intubation might be the cause of statistical heterogeneity of sensitivity and specificity, respectively, for post-extubation airway obstruction.
Conclusions
The cuff leak test has excellent specificity but moderate sensitivity for post-extubation airway obstruction. The high specificity suggests that clinicians should consider intervening in patients with a positive test, but the low sensitivity suggests that patients still need to be closely monitored post-extubation.
Background
Laryngeal edema and airway obstruction following extubation is a major cause of extubation failure [1]. Post-extubation stridor, its clinical sign, has a reported incidence of 2–26% and frequently results in reintubation [1]. Reintubation is associated with an increase in morbidity, duration of mechanical ventilation, and ICU stay [2,3,4,5,6,7,8]. In select cases, systemic corticosteroids before extubation can be used to prevent post-extubation airway complications [9]. Therefore, it is important to estimate the risk of laryngeal edema before extubation.
Since the endotracheal tube precludes direct visualization of the upper airway, the cuff leak test was proposed to predict the presence of laryngeal edema and post-extubation airway obstruction [10, 11]. Theoretically, when there is no laryngeal edema, there is an air leak around the tube after deflating the balloon cuff of the endotracheal tube [12, 13]. In contrast, a failed cuff leak test suggests little or no air leak around the tube, suggesting potential airway obstruction from laryngeal edema [12, 13]. The clinical practice guideline published by the American Thoracic Society and American College of Chest Physicians in 2017 recommends performing a cuff leak test in mechanically ventilated adults who meet extubation criteria to screen those at high risk for post-extubation stridor [14]. This guideline referenced two systematic reviews on the diagnostic accuracy of cuff leak test [15, 16], published in 2009 and 2011. Both reviews demonstrated excellent specificity of the cuff leak test but disagreed with respect to sensitivity. In addition, there have been several studies of the diagnostic accuracy of cuff leak test published after these reviews were completed.
Consequently, we conducted a systematic review and meta-analysis to assess the diagnostic accuracy of the cuff leak test for predicting post-extubation airway obstruction and subsequent reintubation.
Methods
The conduct and reporting of this systematic review followed the PRISMA-DTA Statement [17]. Our review protocol was registered at PROSPERO (CRD42018084357).
We searched Medline, EMBASE, Scopus, ISI Web of Science, and the Cochrane Library for eligible studies from inception to March 16, 2020, without language restrictions [18]. Our search strategy was developed with the help of a medical librarian (Additional file 3: Table S1). We hand-searched the references of included articles for potentially relevant studies.
We examined the diagnostic accuracy of cuff leak test in intubated adult patients awaiting extubation in critical care settings. The index test was cuff leak test regardless of the type of cuff leak test (quantitative or qualitative) and threshold used. The reference standards included post-extubation airway obstruction determined by the original authors and subsequent reintubation. We included observational studies (cross-sectional and cohort studies) that examined the diagnostic accuracy of cuff leak test in critical care settings if: (1) the data were extractable into a 2 × 2 table from the reported data, (2) post-extubation airway obstruction after extubation or reintubation was explicitly reported as the reference standard. We considered both published studies and conference proceedings; however, we included the abstracts from conference proceedings only when they provided data in enough detail to be extractable. We considered interventional studies in critical care settings that examined the efficacy of systemic corticosteroids to prevent post-extubation airway complications; however, we excluded patients to whom systemic corticosteroids were administered after they were judged at high risk of post-extubation airway complications. Two authors (AK and JK) independently screened titles and abstracts obtained from the search and selected potentially relevant articles. Disagreement was resolved through discussion.
The first author (AK) and one of the other authors (JLJ and JK) in duplicate and independently extracted the following data from each study: (1) patient demographics (age, sex); (2) study characteristics (country, study population; duration of mechanical ventilation; mode of ventilation; observation period after extubation); (3) the type of cuff leak test (quantitative or qualitative); (4) numbers of true-positive, false-positive, true-negative, and false-negative; and (5) the reference standards used. Quantitative cuff leak test measures the air leak volume with a cuff deflated and judges the post-extubation airway obstruction based on its absolute volume or proportion in comparison with the expiratory tidal volume against a certain threshold [19]. Qualitative cuff leak test examines the presence or absence of audible expired air around an endotracheal tube, which indicates the pass or failure of the test [19]. In this study, a lack of a cuff leak, having a risk of post-extubation complications, was considered a positive test, while having a leak, suggesting low risk of post-extubation complications was a negative test [19].
Two authors (AK and JLJ) independently assessed the risk of bias using the Quality Assessment for Diagnostic Accuracy Studies-2 (QUADAS-2) tool [20]. Inconsistency was resolved through consensus.
Statistical analysis
We had two primary outcomes: (1) post-extubation airway obstruction and (2) reintubation due to post-extubation airway obstruction. The reference standards for post-extubation airway obstruction included stridor (audible high-pitched inspiratory wheeze) [21], or laryngeal edema defined by the study authors (including confirmation with bronchoscopy [22] or laryngoscope [23]).
We pooled the data using a generalized linear mixed model approach to bivariate random-effects meta-analysis to calculate summary estimates of sensitivity, specificity, and likelihood ratios as well as the associated 95% confidence intervals (CIs) [24]. We pooled prevalence using a random-effects model, with exact binomial estimates of standard deviation and the Freeman–Tukey transformation for zero cells [25]. To examine the sources of heterogeneity, we examined the receiver operating characteristic (ROC) curves, assessed the correlation between sensitivity and specificity, and analyzed whether sensitivity and specificity of cuff leak changed with the type of cuff leak test (qualitative or quantitative), the proportion of women, inclusion or exclusion of reintubated patients in a study, and the length of intubation, using subgroup or meta-regression analysis [14, 19]. We also calculated the sensitivity and specificity of cuff leak test using a cutoff of 110 mL, a value that is frequently used in clinical practice [21]. We tested for publication bias using Deeks’ method [26]. We created a Fagan's nomogram, which determines the posttest probability according to the pretest probability and the calculated positive and negative likelihood ratios [27]. We followed standard diagnostic meta-analytic approaches in focusing on the sensitivity, specificity, and likelihood ratios instead of positive and negative predictive values because predictive values are dependent on the population prevalence of post-extubation complications, which can vary considerably. The threshold of statistical significance was set at P < 0.05. All analyses were performed with Stata SE, version 15.1 (Stata Corp; College Station, TX).
Results
Our literature search produced 2236 studies. After application of inclusion and exclusion criteria, 28 studies involving 4493 extubations were included in the analysis (Fig. 1) [12, 13, 21,22,23, 28,29,30,31,32,33,34,35,36,37,38,39,40,41,42,43,44,45,46,47,48,49,50]. Among the 28 included studies, 27 were published in English and one in Korean [35].
Twenty-seven studies (96%) were prospective [12, 13, 21,22,23, 28,29,30,31,32,33,34,35,36,37, 39,40,41,42,43,44,45,46,47,48,49,50]. Nine studies and ten were conducted in medical [21, 22, 31, 32, 35,36,37, 43, 49] and mixed intensive care units [12, 13, 23, 28, 33, 34, 40, 41, 44, 45], respectively (Table 1). The included studies were published between 1992 and 2019. All but one were published studies and the remaining one an abstract in a conference proceeding. Eight studies were conducted in the USA, four each in France and Taiwan, three in Thailand, two in Egypt, one each in Australia, Belgium, India, Iran, Italy, South Korea, and Turkey. The median sample size was 101, ranging 34–543 (interquartile range 51–236). The reported mean/median duration of mechanical ventilation ranged from 2 to 28.1 days.
Five studies used a qualitative cuff leak test (auscultation of airflow) [23, 28, 31, 39, 45], and 21 used a quantitative measurement of the cuff leak [12, 13, 21, 22, 29, 30, 32,33,34,35,36,37, 40,41,42,43,44, 46, 47, 49, 50] (Table 2). One study reported the results from both qualitative and quantitative cuff leak tests examined in a single cohort [48]. The remaining study reported on the data of patients who underwent either qualitative or quantitative cuff leak test [38]. The most frequent cutoff values for quantitative cuff leak tests ranged from 50 to 283 mL (median, 110 mL) in volume and from 10 to 57% in proportion. Twenty studies used assist control ventilation [12, 13, 21, 22, 29, 30, 32,33,34,35,36,37,38, 41,42,43,44, 46, 47, 49] , and four used spontaneous breathing including pressure support ventilation [23, 31, 40, 45]. Another study examined either of these two ventilation modes [28]. One study applied the ambu bag, while the cuff leak was tested [39]. One study performed a qualitative cuff leak test during spontaneous breathing via T-tube and during cough while still intubated [48]. The remaining one did not report the mode of ventilation used [50]. Twenty-four studies used stridor [12, 13, 21, 28,29,30,31,32,33,34,35,36,37,38,39, 41,42,43, 45,46,47,48,49,50] and three used direct visualization of the airway following extubation as reference standards [22, 23, 40] for airway obstruction. One study used either of these reference standards [44].
Three of the 28 studies were at low risk of bias for all QUADAS-2 risk of bias domains (Table 3). Seventeen studies (60.7%) were deemed at low risk of bias for the domain of patient selection. Eight out of 22 studies that assessed quantitative cuff leak prespecified the cutoff of cuff leak; fourteen studies (50%) were deemed to have adequately assessed the index test. A reference standard was adequately assessed in 18 studies (64.3%). Study participants were adequately followed up in 19 studies (67.9%).
Post-extubation airway obstruction
The prevalence of post-extubation airway obstruction ranged from 4 to 37% (pooled estimate 9%; 95% CI 7–11; I2 = 86.3%). The pooled sensitivity and specificity of cuff leak test for post-extubation airway obstruction were 0.62 (95% CI 0.49–0.73; I2 = 81.6%) and 0.87 (95% CI 0.82–0.90; I2 = 97.8%), respectively. The forest plots are shown in Additional file 1: Figure S1. The pooled positive and negative likelihood ratios were 4.63 (95% CI 3.44–6.22) and 0.44 (95% CI 0.32–0.60), respectively. The area under the summary receiver operating characteristic (SROC) curve was 0.85 (95% CI 0.82–0.88) (Fig. 2), and the pooled diagnostic odds ratio (DOR) was 10.54 (95% CI 6.26–17.76) (Additional file 4: Table S2). There was no evidence of publication bias (p = 0.189).
Summary receiver operating characteristics (SROC) curve of the cuff leak test for predicting post-extubation upper airway obstruction. The circle size represents the number of extubation in each study. The pooled sensitivity is 0.62 (95% CI 0.49–0.73) and the pooled specificity is 0.87 (95% CI 0.82–0.90), with the area under the SROC curve of 0.85 (95% CI 0.82–0.88). The prediction region illustrates the extent of statistical heterogeneity by depicting a region within there is 95% confidence that the true sensitivity and specificity of a future study should lie
Subgroup analysis suggested that the specificity was similar between the qualitative and quantitative cuff leak tests (qualitative 0.89 [95% CI 0.82–0.96] vs. quantitative 0.86 [95% CI 0.81–0.91], p < 0.01). While the sensitivity difference was clinically important, it was statistically not significant (quantitative 0.67 [95% CI 0.56–0.78] vs. qualitative 0.35 [95% CI 0.12–0.57], p = 0.07). The specificity of the cuff leak test was slightly but significantly (p = 0.02) worse in patients intubated more than 6 days (0.85 [95% CI 0.76–0.95]) than those intubated ≤ 6 days (0.87 [95% CI 0.81–0.92]), with no significant difference in the sensitivity. There was no difference of sensitivity or specificity with the proportion of women or with the exclusion of reintubated patients. The sensitivity and specificity of cuff leak test based on a cut point of 110 mL were 0.44 (95% CI 0.31–0.59) and 0.91 (95% CI 0.82–0.95), respectively.
A nomogram based on the pretest probability of 9% (the incidence of stridor in the studies included in our study) is provided (Fig. 3).
Reintubation
The prevalence of reintubation varied from 0 to 11% (pooled estimate: 3%; 95% CI 1–5; I2 = 79%). The pooled sensitivity and specificity of the cuff leak test for reintubation were 0.66 (95% CI 0.46–0.81; I2 = 48.9%) and 0.88 (95% CI 0.83–0.92; I2 = 87.4%), respectively. The forest plots of the sensitivity and specificity of the cuff leak test for predicting reintubation are shown in Additional file 2: Figure S2. The pooled positive and negative likelihood ratios were 5.59 (95% CI 3.48–8.98) and 0.39 (95% CI 0.23–0.66). The area under the SROC curve was 0.88 (95% CI 0.83–0.90) (Fig. 4) and the pooled DOR was 14.34 (95% CI 5.65–36.42) (Additional file 4: Table S2). There was no evidence of publication bias (p = 0.52).
Summary receiver operating characteristics (SROC) curve of the cuff leak test for predicting reintubation. The circle size represents the number of extubation in each study. The pooled sensitivity is 0.66 (95% CI 0.46–0.81) and the pooled specificity is 0.88 (95% CI 0.83–0.92), with the area under the SROC curve of 0.88 (95% CI 0.83–0.90). The prediction region illustrates the extent of statistical heterogeneity by depicting a region within there is 95% confidence that the true sensitivity and specificity of a future study should lie
Discussion
Our study found that the cuff leak test has excellent specificity but moderate sensitivity for post-extubation airway obstruction. The cuff leak test thus works better to rule in than to rule out potential post-extubation airway obstruction. However, the false-negative rate of 38% suggests that the cuff leak test may fail to identify some patients with post-extubation airway obstruction.
Our study found that the specificity of the cuff leak test for post-extubation airway obstruction was excellent, which is consistent with two previous systematic reviews [15, 16]. In contrast, Ochoa et al. and Zhou et al. concluded that the sensitivity of cuff leak testing for post-extubation airway obstruction was 56% and 80%, respectively [15, 16]: Our pooled sensitivity was 62%, which fell between those two findings. We included nearly double the number of studies that their reviews did. Furthermore, the additional studies we included were higher quality, potentially making our findings more reliable.
Our analysis found that the qualitative cuff leak test had low sensitivity (35%) in predicting post-extubation airway obstruction. This has been consistently found in recent studies. The most likely explanation is the subjective nature of this test. In addition, since Schnell et al. provided data from three different methods of qualitative cuff leak testing [48], the sensitivity of which were all around 30%; this study may have been overweighed, although repeat analysis limiting Schnell’s study to a single data contribution did not change the sensitivity of the qualitative test. In contrast, the specificity of both qualitative and quantitative cuff leak tests was high, nearly 90%, while there was a statistically significant difference between two methods, clinically both performed equally well. A cutoff of 110 mL also had a low sensitivity (44%) and high specificity for predicting post-extubation airway obstruction. We thus conclude that the cuff leak test has high specificity and can be used to select patients to consider treating with systemic corticosteroids, but its low sensitivity suggests that the traditional practice of closely observing all patients in the immediate post-extubation period should be continued. Consistent with these findings, the nomogram suggested that while a negative cuff leak test represents low possibility of post-extubation airway obstruction, a positive test still provides a relatively low posttest probability.
The guideline by ATS/ACCP provided a conditional recommendation regarding cuff leak test [14], because failing the cuff leak test might lead a delay in extubation and an increase in complications such as barotrauma and ventilator-associated pneumonia. The guideline weakly recommended that the cuff leak test be reserved for high-risk patients, who experienced a traumatic intubation, were intubated more than 6 days, have a large endotracheal tube, are female, or were reintubated after an unplanned extubation [14]. Our analysis found that the length of intubation had a small impact on the specificity of cuff leak test. Female sex and reintubation had no impact of the accuracy of cuff leak test. Since the original studies included in our review examined non-selected patients with respect to the risk of post-extubation airway obstruction and the sensitivity of cuff leak test is moderate, we support the idea of the ATS/ACCP guideline to reserve cuff leak test for high-risk patients.
Our study suggested that the cuff leak test has moderate sensitivity and excellent specificity for reintubation. Although the sensitivity in our study was similar to those of previous meta-analyses, the specificity in our study was slightly higher [15, 16]. The area under the SROC curve and DOR were also greater than previously reported [16]. Thus, a failed cuff leak test may serve as a good marker for those at risk of reintubation, if post-extubation airway obstruction is not treated adequately.
The limitation of cuff leak test has been repeatedly discussed. Cuff leak test can be susceptible to relationship of tube size to laryngeal diameter [41], respiratory system compliance and resistance, inspiratory flow, expiatory flow and time, and airway collapse [51], and clinicians should bear in mind that the ability of cuff leak test may vary according to the condition or type of patients [52]. Additionally, coughing during cuff deflation test hinders accurate measurement of the leak volume and lowers the reproducibility. A previous physiological study suggested that while patients were sedated and paralyzed, the cuff leak volume was reliably measurable [53]. An adequate amount of sedatives and opioids can suppress coughing during the airway suctioning before cuff leak test or cuff deflation during the test. Further, cuff leak testing is recommended several hours before extubation, which allows the arousal of patients from sedation by the time of extubation. Thus, we may be able to at least attempt to increase the reliability of cuff leak measurement.
Few tests are available to estimate the risk of post-extubation airway complications. A case series of three patients suggested that video laryngoscopy enabled visualization of laryngeal edema prior to extubation [54], but its clinical efficacy in estimating post-extubation airway complications is yet to be determined. Several studies have examined the role of laryngeal ultrasonography in adult patients. Laryngeal air column width difference is the difference between width of airway at the level of the vocal cord with cuff inflated and deflated. Its reported sensitivity and specificity varied across studies, ranging from 50 to 91% and 54 to 72%, respectively [44, 46, 49]. Laryngeal air column width ratio is the ratio of air column width before extubation over that after intubation. It has been examined in only one study [55] and needs further validations. Thus, no single available options can correctly estimate the risk of post-extubation airway complications. Clinicians should not overly rely on one single test in predicting the success or failure of extubation.
Our study had several strengths and limitations. Strengths included a comprehensive search in five databases without language restrictions. This allowed us to conduct relevant subgroup analyses with a larger number of studies. Further, inclusion of non-English studies facilitates the generalizability of our findings in various clinical settings [56].
Our study had some limitations. First, the definition of post-extubation airway obstruction differed across studies. Stridor was more frequently used as the reference standards than laryngeal edema (as assessed with endoscopy). Laryngeal edema may be more frequent than stridor, because stridor and respiratory distress occur when laryngeal edema narrows the airway by ≥ 50% [57]. However, laryngeal edema is not always screened for in extubated patients, and the presence of stridor is an accepted sign of respiratory distress that triggers a concern for airway obstruction. Thus, the finding of our study is generalizable to the clinical practice. Second, although we attempted to include in the analysis the incidence of stridor due to post-extubation airway events, some patients might have had a concurrent clinical state that necessitated high minute ventilation or tachypnea through an edematous airway, which manifested as ‘stridor.’ Therefore, we might not have been able to completely separate stridor due to post-extubation airway events from stridor due to other etiologies, such as respiratory insufficiency. This limitation also applies to reintubation. Third, whether to reintubate patients is subject to treating physicians’ discretion and the effect of treatment to abort post-extubation stridor. Therefore, the value of cuff leak test in predicting the need for reintubation in clinical practice may be limited along with the third limitation. However, prevention of post-extubation airway obstruction is more important than reintubation per se. Once the cuff leak test identifies patients at high risk of post-extubation airway obstruction, prophylactic systemic corticosteroids are indicated [9, 14, 58]. Fourth, 15 out of 23 studies that assessed the quantitative cuff leak test determined the cutoff with the knowledge of the results of the reference standards. It is known that data-driven optimization of the cutoff can lead to overestimation of test performance [59]. Thus, the pooled accuracy of quantitative cuff leak test in our study can be an overestimation; the optimal cutoff is still unknown. Fifth, the quantitative cutoff for a positive test varied between the studies. Since we had aggregate data from each included study, we failed to determine the optimal cutoff of cuff leak test. Finally, there was substantial statistical heterogeneity in the pooled sensitivity and specificity for both outcomes. The presence of statistical heterogeneity is a common issue intrinsic to the meta-analysis of diagnostic accuracy of a test, given the clinical and methodological diversity in original studies as well as the possible relationship between sensitivity and specificity, as exemplified in ROC curves in which more sensitive cut points have lower specificity (and vice versa). Our subgroup analyses suggested that the type of cuff leak test (quantitative versus qualitative) and length of intubation might have been the cause of statistical heterogeneity of sensitivity and specificity, respectively, for post-extubation airway obstruction.
Conclusion
The cuff leak test has excellent specificity but moderate sensitivity for post-extubation airway obstruction. The cuff leak test is a useful tool in the decision-making about extubation, but the low sensitivity suggests that a negative test cannot completely exclude post-extubation airway obstruction and that patients still need to be closely monitored post-extubation. The higher specificity suggests that clinicians should consider intervening in patients with systemic corticosteroids in response to a positive test. Continued research to find better modalities to rule out post-extubation airway obstruction is needed.
Availability of data and materials
All data generated or analyzed during this study are included in this published article and its supplementary information files.
References
Pluijms WA, van Mook WN, Wittekamp BH, Bergmans DC. Postextubation laryngeal edema and stridor resulting in respiratory failure in critically ill adult patients: updated review. Crit Care. 2015;19:295.
Torres A, Gatell JM, Aznar E, el-Ebiary M, Puig de la Bellacasa J, Gonzalez J, Ferrer M, Rodriguez-Roisin R. Re-intubation increases the risk of nosocomial pneumonia in patients needing mechanical ventilation. Am J Respir Crit Care Med. 1995;152(1):137–41.
Frutos-Vivar F, Esteban A, Apezteguia C, Gonzalez M, Arabi Y, Restrepo MI, Gordo F, Santos C, Alhashemi JA, Perez F, et al. Outcome of reintubated patients after scheduled extubation. J Crit Care. 2011;26(5):502–9.
Epstein SK, Ciubotaru RL. Independent effects of etiology of failure and time to reintubation on outcome for patients failing extubation. Am J Respir Crit Care Med. 1998;158(2):489–93.
Epstein SK, Ciubotaru RL, Wong JB. Effect of failed extubation on the outcome of mechanical ventilation. Chest. 1997;112(1):186–92.
Seymour CW, Martinez A, Christie JD, Fuchs BD. The outcome of extubation failure in a community hospital intensive care unit: a cohort study. Crit Care. 2004;8(5):R322-327.
Gowardman JR, Huntington D, Whiting J. The effect of extubation failure on outcome in a multidisciplinary Australian intensive care unit. Crit Care Resusc. 2006;8(4):328–33.
Thille AW, Harrois A, Schortgen F, Brun-Buisson C, Brochard L. Outcomes of extubation failure in medical intensive care unit patients. Crit Care Med. 2011;39(12):2612–8.
Kuriyama A, Umakoshi N, Sun R. Prophylactic corticosteroids for prevention of postextubation stridor and reintubation in adults: a systematic review and meta-analysis. Chest. 2017;151(5):1002–10.
Adderley RJ, Mullins GC. When to extubate the croup patient: the ‘leak’ test. Can J Anesth. 1987;34(3):304–6.
Potgieter PD, Hammond JM. “Cuff” test for safe extubation following laryngeal edema. Crit Care Med. 1988;16(8):818.
De Bast Y, De Backer D, Moraine JJ, Lemaire M, Vandenborght C, Vincent JL. The cuff leak test to predict failure of tracheal extubation for laryngeal edema. Intensive Care Med. 2002;28(9):1267–72.
Jaber S, Chanques G, Matecki S, Ramonatxo M, Vergne C, Souche B, Perrigault PF, Eledjam JJ. Post-extubation stridor in intensive care unit patients. Risk factors evaluation and importance of the cuff-leak test. Intensive Care Med. 2003;29(1):69–74.
Girard TD, Alhazzani W, Kress JP, Ouellette DR, Schmidt GA, Truwit JD, Burns SM, Epstein SK, Esteban A, Fan E, et al. An Official American Thoracic Society/American College of Chest Physicians clinical practice guideline: liberation from mechanical ventilation in critically ill adults. Rehabilitation protocols, ventilator liberation protocols, and cuff leak tests. Am J Respir Crit Care Med. 2017;195(1):120–33.
Ochoa ME, Marin Mdel C, Frutos-Vivar F, Gordo F, Latour-Perez J, Calvo E, Esteban A. Cuff-leak test for the diagnosis of upper airway obstruction in adults: a systematic review and meta-analysis. Intensive Care Med. 2009;35(7):1171–9.
Zhou T, Zhang HP, Chen WW, Xiong ZY, Fan T, Fu JJ, Wang L, Wang G. Cuff-leak test for predicting postextubation airway complications: a systematic review. J Evid Based Med. 2011;4(4):242–54.
McInnes MDF, Moher D, Thombs BD, McGrath TA, Bossuyt PM, Clifford T, Cohen JF, Deeks JJ, Gatsonis C, et al. Preferred reporting items for a systematic review and meta-analysis of diagnostic test accuracy studies: the PRISMA-DTA statement. JAMA. 2018;319(4):388–96.
Jackson JL, Kuriyama A, Anton A, Choi A, Fournier JP, Geier AK, Jacquerioz F, Kogan D, Scholcoff C, Sun R. The accuracy of google translate for abstracting data from non-English-language trials for systematic reviews. Ann Internal Med. 2019;171:677.
Argalious MY. The cuff leak test: does it “leak” any information? Respir Care. 2012;57(12):2136–7.
Whiting PF, Rutjes AW, Westwood ME, Mallett S, Deeks JJ, Reitsma JB, Leeflang MM, Sterne JA, Bossuyt PM. Group Q-: QUADAS-2: a revised tool for the quality assessment of diagnostic accuracy studies. Ann Intern Med. 2011;155(8):529–36.
Miller RL, Cole RP. Association between reduced cuff leak volume and postextubation stridor. Chest. 1996;110(4):1035–40.
Chung Y-H, Chao T-Y, Chiu C-T, Lin M-C. The cuff-leak test is a simple tool to verify severe laryngeal edema in patients undergoing long-term mechanical ventilation. Crit Care Med. 2006;34(2):409–14.
Fisher MM, Raper RF. The “cuff-leak” test for extubation. Anaesthesia. 1992;47(1):10–2.
Chu H, Cole SR. Bivariate meta-analysis of sensitivity and specificity with sparse data: a generalized linear mixed model approach. J Clin Epidemiol. 2006;59(12):1331–2 (author reply 1332–3).
Freeman MF, Tukey JW. Transformations related to the angular and the square root. Ann Math Stat. 1950;21:607–11.
Deeks JJ, Macaskill P, Irwig L. The performance of tests of publication bias and other sample size effects in systematic reviews of diagnostic test accuracy was assessed. J Clin Epidemiol. 2005;58(9):882–93.
Deeks JJ, Altman DG. Diagnostic tests 4: likelihood ratios. BMJ (Clin Res Ed). 2004;329(7458):168–9.
Marik PE. The cuff-leak test as a predictor of postextubation stridor: a prospective study. Respir Care. 1996;41(6):509–11.
Engoren M. Evaluation of the cuff-leak test in a cardiac surgery population. Chest. 1999;116(4):1029–31.
Sandhu RS, Pasquale MD, Miller K, Wasser TE. Measurement of endotracheal tube cuff leak to predict postextubation stridor and need for reintubation. J Am Coll Surg. 2000;190(6):682–7.
Maury E, Guglielminotti J, Alzieu M, Qureshi T, Guidet B, Offenstadt G. How to identify patients with no risk for postextubation stridor? J Crit Care. 2004;19(1):23–8.
Erginel S, Ucgun I, Yildirim H, Metintas M, Parspour S. High body mass index and long duration of intubation increase post-extubation stridor in patients with mechanical ventilation. Tohoku J Exp Med. 2005;207(2):125–32.
Kriner EJ, Shafazand S, Colice GL. The endotracheal tube cuff-leak test as a predictor for postextubation stridor. Respir Care. 2005;50(12):1632–8.
Cheng KC, Hou CC, Huang HC, Lin SC, Zhang H. Intravenous injection of methylprednisolone reduces the incidence of postextubation stridor in intensive care unit patients. Crit Care Med. 2006;34(5):1345–50.
Lim SY, Suh GY, Kyung SY, An CH, Lee SP, Park JW, Jeong SH, Ham HS, Ahn YM, Lim SY, et al. Risk factors of extubation failure and analysis of cuff leak test as a predictor for postextubation stridor. Tuberc Respir Dis. 2006;61(1):34–40.
Lee CH, Peng MJ, Wu CL. Dexamethasone to prevent postextubation airway obstruction in adults: a prospective, randomized, double-blind, placebo-controlled study. Crit Care. 2007;11(4):R72.
Wang CL, Tsai YH, Huang CC, Wu YK, Ye MZ, Chou HM, Shu SC, Lin MC. The role of the cuff leak test in predicting the effects of corticosteroid treatment on postextubation stridor. Chang Gung Med J. 2007;30(1):53–61.
Shin SH, Heath K, Reed S, Collins J, Weireter LJ, Britt LD. The cuff leak test is not predictive of successful extubation. Am Surg. 2008;74(12):1182–5.
Sukhupanyarak S. Risk factors evaluation and the cuff leak test as predictors for postextubation stridor. J Med Assoc Thailand. 2008;91(5):648–53.
Antonaglia V, Vergolini A, Pascotto S, Bonini P, Renco M, Peratoner A, Buscema G, De Simoni L. Cuff-leak test predicts the severity of postextubation acute laryngeal lesions: a preliminary study. Eur J Anaesthesiol. 2010;27(6):534–41.
Gros A, Holzapfel L, Marque S, Perard L, Demingeon G, Piralla B, Gaillard S, Tchenio X. Intra-individual variation of the cuff-leak test as a predictor of post-extubation stridor. Respir Care. 2012;57(12):2026–31.
Keeratichananont W, Limthong T, Keeratichananont S. Cuff leak volume as a clinical predictor for identifying post-extubation stridor. J Med Assoc Thailand. 2012;95(6):752–5.
Radhi S, Guerra D, Alalawi R, Raj R, Nugent K. Cuff leak tests at the time of extubation correlate with voice quality assessment. ICU Dir. 2012;3(1):27–30.
Sutherasan Y, Theerawit P, Hongphanut T, Kiatboonsri C, Kiatboonsri S. Predicting laryngeal edema in intubated patients by portable intensive care unit ultrasound. J Crit Care. 2013;28(5):675–80.
Patel AB, Ani C, Feeney C. Cuff leak test and laryngeal survey for predicting post-extubation stridor. Indian J Anaesth. 2015;59(2):96–102.
El-Baradey GF, El-Shmaa NS, Elsharawy F. Ultrasound-guided laryngeal air column width difference and the cuff leak volume in predicting the effectiveness of steroid therapy on postextubation stridor in adult. Are they useful? J Crit Care. 2016;36:272–6.
Sahbal MA, Mohamed KA, Zaghla HH, Kenawy MM. Laryngeal ultrasound versus cuff leak test in prediction of post-extubation stridor. Egypt J Crit Care Med. 2017;5(3):83–6.
Schnell D, Planquette B, Berger A, Merceron S, Mayaux J, Strasbach L, Legriel S, Valade S, Darmon M, Meziani F. Cuff leak test for the diagnosis of post-extubation stridor. J Intensive Care Med. 2017;34:391–6.
Mikaeili H, Yazdchi M, Tarzamni MK, Ansarin K, Ghasemzadeh M. Laryngeal ultrasonography versus cuff leak test in predicting postextubation stridor. J Cardiovasc Thorac Res. 2014;6(1):25–8.
Samanta S, Azim A, Baronia A, Poddar B, Gurjar M, Singh R: Clinical risk, cuff leak test and laryngeal ultrasound based scoring system to predict postextubation stridor in intensive care unit patients. Eur Respir J 2019, 54.
Prinianakis G, Alexopoulou C, Mamidakis E, Kondili E, Georgopoulos D. Determinants of the cuff-leak test: a physiological study. Crit Care. 2005;9(1):R24-31.
De Backer D. The cuff-leak test: what are we measuring? Crit Care. 2005;9(1):31–3.
Pettignano R, Holloway SE, Hyman D, LaBuz M. Is the leak test reproducible? South Med J. 2000;93(7):683–5.
Newmark JL, Ahn YK, Adams MC, Bittner EA, Wilcox SR. Use of video laryngoscopy and camera phones to communicate progression of laryngeal edema in assessing for extubation: a case series. J Intensive Care Med. 2013;28(1):67–71.
Venkategowda PM, Mahendrakar K, Rao SM, Mutkule DP, Shirodkar CG, Yogesh H. Laryngeal air column width ratio in predicting post extubation stridor. Indian J Crit Care Med. 2015;19(3):170–3.
Jackson JL, Kuriyama A. How often do systematic reviews exclude articles not published in English? J Gen Intern Med. 2019;34(8):1388–9.
Mackle T, Meaney J, Timon C. Tracheoesophageal compression associated with substernal goitre. Correlation of symptoms with cross-sectional imaging findings. J Laryngol Otol. 2007;121(4):358–61.
Kuriyama A, Egawa S, Kataoka J, Sakuraya M, Matsumura M. Adverse events associated with prophylactic corticosteroid use before extubation: a cohort study. Ann Transl Med. 2020;8(14):853.
Leeflang MM, Moons KG, Reitsma JB, Zwinderman AH. Bias in sensitivity and specificity caused by data-driven selection of optimal cutoff values: mechanisms, magnitude, and solutions. Clin Chem. 2008;54(4):729–37.
Acknowledgements
We would like to sincerely thank Ms. Elizabeth Suelzer, MLIS, AHIP for her help with literature search.
Funding
There was no funding source for this study.
Author information
Authors and Affiliations
Contributions
AK and JLJ substantially contributed to conception of the study design, data acquisition, data analysis, interpretation, and the writing and critical revision of the manuscript. JK substantially contributed to data acquisition, interpretation, and critical revision of the manuscript. All authors read and approved the submission of the final manuscript.
Corresponding author
Ethics declarations
Ethics approval and consent to participate
This study is a systematic review and an ethics approval was not necessary.
Consent for publication
Not applicable.
Competing interests
The authors declare that they have no competing interests.
Additional information
Publisher's Note
Springer Nature remains neutral with regard to jurisdictional claims in published maps and institutional affiliations.
Supplementary information
Additional file 1: Figure S1.
Forest plot of the sensitivity and specificity of the cuff leak test for predicting post-extubation upper airway obstruction.
Additional file 2: Figure S2.
Forest plot of the sensitivity and specificity of the cuff leak test for predicting reintubation.
Additional file 3: Table S1.
Search strategy.
Additional file 4: Table S2.
The pooled diagnostic accuracy of cuff leak test for post-extubation airway obstruction and reintubation.
Rights and permissions
Open Access This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.
About this article
Cite this article
Kuriyama, A., Jackson, J.L. & Kamei, J. Performance of the cuff leak test in adults in predicting post-extubation airway complications: a systematic review and meta-analysis. Crit Care 24, 640 (2020). https://doi.org/10.1186/s13054-020-03358-8
Received:
Accepted:
Published:
DOI: https://doi.org/10.1186/s13054-020-03358-8