Skip to main content

Angiotensin in ECMO patients with refractory shock

A Letter to this article was published on 26 April 2019

Refractory vasodilation and catecholamine resistance are common in septic shock. Changes in receptor signaling, excessive production of nitric oxide, and absolute or relative deficiencies of vasoactive hormones, including cortisol, vasopressin, and angiotensin II, play a role. Angiotensin II (Ang II) was previously available as a vasopressor but removed from the market in the 1990s. Interest was re-ignited following the Angiotensin II for the Treatment of Vasodilatory Shock (ATHOS-3) study, a randomized controlled trial in patients with refractory shock which confirmed that Ang II was effective at maintaining mean arterial pressure and reducing norepinephrine requirements without an increase in side effects [1]. Patients receiving renal replacement therapy also had improved survival and faster recovery of renal function [2]. Recent literature noted the potential role of Ang II in other types of shock [3].

The major physiological effects of Ang II relate to maintenance of hemodynamic stability and fluid and electrolyte regulation (Table 1). Angiotensinogen, the precursor of angiotensin, is produced primarily by the liver and released into the systemic circulation where it is converted to angiotensin I (Ang I). Ang I is cleaved into Ang II, predominantly by angiotensin converting enzyme (ACE), an endothelium bound protein that is primarily expressed in the pulmonary and renal capillary beds. In patients with acute respiratory distress syndrome, ACE insufficiency has been reported [4]. In veno-arterial ECMO, a proportion of blood bypasses the lungs, which further limits the conversion of Ang I to Ang II. Other conditions associated with reduced Ang II levels include Gram-negative sepsis where endotoxinemia can deactivate ACE. Importantly, low levels of Ang II and ACE are associated with increased mortality [5].

Table 1 Main physiological effects of angiotensin II

We report the successful management of seven patients (four male; mean age 36 years) with severe cardiorespiratory failure and refractory shock treated with extracorporeal membrane oxygenation (ECMO) who received Ang II in the context of the ATHOS-3 trial [1] or a compassionate use program (Table 2). Following initiation of Ang II, a profound effect on blood pressure was seen and the doses of vasopressors were reduced quickly. Time to cessation of vasopressors and catecholamines ranged from 16 h to 8 days. Six patients were discharged home alive.

Table 2 Patient characteristics

In conclusion, in patients with severe cardio-respiratory failure requiring ECMO, treatment with Ang II in addition to standard supportive care enabled rapid decatecholaminization. Underlying ACE deficiency may be a contributing factor. Further studies are necessary to confirm the findings.

Abbreviations

ACE:

Angiotensin converting enzyme

Ang I:

Angiotensin I

Ang II:

Angiotensin II

ATHOS:

Angiotensin II for the Treatment of Vasodilatory Shock

ECMO:

Extracorporeal membrane oxygenation

References

  1. Khanna A, English SW, Wang XS, et al. Angiotensin II for the treatment of vasodilatory shock. N Engl J Med. 2017;377(5):419–30.

    Article  CAS  Google Scholar 

  2. Tumlin JA, Murugan R, Deane AM, et al. Outcome in patients with vasodilatory shock and renal replacement therapy treated with intravenous angiotensin II. Crit Care Med. 2018;46(6):949–57.

    Article  CAS  Google Scholar 

  3. Busse LW, McCurdy MT, Ali O, Hall A, Chen H, Ostermann M. The effect of angiotensin II on blood pressure in patients with circulatory shock: a structured review of the literature. Crit Care. 2017;21(1):324.

    Article  Google Scholar 

  4. Orfanos SE, Armaganidis A, Glynos C, et al. Pulmonary capillary endothelium-bound angiotensin-converting enzyme activity in acute lung injury. Circulation. 2000;102(16):2011–8.

    Article  CAS  Google Scholar 

  5. Zhang W, Chen X, Huang L, et al. Severe sepsis: Low expression of the renin-angiotensin system is associated with poor prognosis. Exp Ther Med. 2014;7(5):1342–8.

    Article  CAS  Google Scholar 

Download references

Acknowledgements

The authors would like to thank the patients for allowing the publication of their anonymized data and contributing to the dissemination of information. We are also grateful to the research nurses and coordinators who helped with the successful conduct of the ATHOS -3 study.

Funding

Not applicable.

Availability of data and materials

The datasets used and analyzed during the current study are available from the corresponding author on reasonable request.

Author information

Authors and Affiliations

Authors

Contributions

MO wrote the first draft of the paper and DWB, MDH, and KG revised the draft. GWL is a clinical fellow who helped with data collection. All authors reviewed the drafts, provided input, and approved the final version.

Corresponding author

Correspondence to Marlies Ostermann.

Ethics declarations

Ethics approval and consent to participate

The case series includes patients who participated in the ATHOS-3 study or received angiotensin II in the context of a compassionate treatment program. The ATHOS-3 study was fully approved by an independent research ethics committee.

Consent for publication

As part of the ATHOS-3 study, patients gave consent for their data and results to be published in an anonymized format.

Competing interests

The authors declare that they have no competing interests.

Publisher’s Note

Springer Nature remains neutral with regard to jurisdictional claims in published maps and institutional affiliations.

Rights and permissions

Open Access This article is distributed under the terms of the Creative Commons Attribution 4.0 International License (http://creativecommons.org/licenses/by/4.0/), which permits unrestricted use, distribution, and reproduction in any medium, provided you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons license, and indicate if changes were made. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated.

Reprints and Permissions

About this article

Check for updates. Verify currency and authenticity via CrossMark

Cite this article

Ostermann, M., Boldt, D.W., Harper, M.D. et al. Angiotensin in ECMO patients with refractory shock. Crit Care 22, 288 (2018). https://doi.org/10.1186/s13054-018-2225-4

Download citation

  • Received:

  • Accepted:

  • Published:

  • DOI: https://doi.org/10.1186/s13054-018-2225-4