Effect of norepinephrine dosage and calibration frequency on accuracy of pulse contour-derived cardiac output

Abstract

Introduction

Continuous cardiac output monitoring is used for early detection of hemodynamic instability and guidance of therapy in critically ill patients. Recently, the accuracy of pulse contour-derived cardiac output (PCCO) has been questioned in different clinical situations. In this study, we examined agreement between PCCO and transcardiopulmonary thermodilution cardiac output (COTCP) in critically ill patients, with special emphasis on norepinephrine (NE) administration and the time interval between calibrations.

Methods

This prospective, observational study was performed with a sample of 73 patients (mean age, 63 ± 13 years) requiring invasive hemodynamic monitoring on a non-cardiac surgery intensive care unit. PCCO was recorded immediately before calibration by COTCP. Bland-Altman analysis was performed on data subsets comparing agreement between PCCO and COTCP according to NE dosage and the time interval between calibrations up to 24 hours. Further, central artery stiffness was calculated on the basis of the pulse pressure to stroke volume relationship.

Results

A total of 330 data pairs were analyzed. For all data pairs, the mean COTCP (±SD) was 8.2 ± 2.0 L/min. PCCO had a mean bias of 0.16 L/min with limits of agreement of -2.81 to 3.15 L/min (percentage error, 38%) when compared to COTCP. Whereas the bias between PCCO and COTCP was not significantly different between NE dosage categories or categories of time elapsed between calibrations, interchangeability (percentage error <30%) between methods was present only in the high NE dosage subgroup (≥0.1 μg/kg/min), as the percentage errors were 40%, 47% and 28% in the no NE, NE < 0.1 and NE ≥ 0.1 μg/kg/min subgroups, respectively. PCCO was not interchangeable with COTCP in subgroups of different calibration intervals. The high NE dosage group showed significantly increased central artery stiffness.

Conclusions

This study shows that NE dosage, but not the time interval between calibrations, has an impact on the agreement between PCCO and COTCP. Only in the measurements with high NE dosage (representing the minority of measurements) was PCCO interchangeable with COTCP.

Introduction

Cardiac output (CO) monitoring in high-risk patients has gained increasing interest because early detection of hemodynamic instability can reduce morbidity in these patients [13]. Investigators in several studies evaluating goal-directed protocols have reported improved outcomes due to immediate treatment to prevent or resolve organ ischemia [4, 5]. The PiCCOplus system (Pulsion Medical Systems, Munich, Germany) allows continuous CO measurement by pulse contour analysis (PCCO). Calibration of PCCO is performed by intermittent transcardiopulmonary thermodilution cardiac output (COTCP). It has been demonstrated that PCCO agrees with pulmonary artery thermodilution CO [68] and with COTCP [9, 10] in cardiac surgery patients. However, the reliability of PCCO has been questioned in clinical scenarios such as acute hemorrhage and subsequent norepinephrine (NE) administration [11], changes in vascular tone [12], increased intra-abdominal pressure [13] or time interval between calibrations [14]. Therefore, the clinician needs to consider these confounders when interpreting PCCO values and prompting therapeutic decisions.

The present prospective observational study investigated a large group of critically ill patients with regard to whether agreement between PCCO and COTCP is affected by different NE dosages or by the time interval between calibrations. On the basis of the existing literature, we generated the following two hypotheses: (1) Increasing NE dosage results in decreased agreement between PCCO and COTCP, and (2) increasing the time interval between calibrations of PCCO results in decreased agreement between PCCO and COTCP.

Only rare data are available about the usage of PCCO calibrations in clinical practice. Therefore, we retrospectively evaluated whether NE dosage or severity of disease as measured by the Acute Physiology and Chronic Health Evaluation II score (APACHE II score) had an influence on calibration frequency on our intensive care unit (ICU).

Materials and methods

Patients

In this prospective observational study, critically ill patients equipped with invasive hemodynamic monitoring by the PiCCOplus system (version 6.0) on our noncardiac ICU between September 2007 and July 2008 were included. The study was approved by our institutional review board in compliance with the Helsinki Declaration (Ethics Committee of the University Hospital Schleswig-Holstein, Campus Kiel, Kiel, Germany). Patients and/or relatives gave their informed consent for the patients' data to be used in the analysis. Invasive hemodynamic monitoring was performed according to the judgment of the attending physician on the ICU. Exclusion criteria were cardiac arrhythmias, a permanent pacemaker or any other mechanical cardiac support and known valvular heart disease.

Hemodynamic measurements

In all patients, a central venous catheter and a thermistor-tipped arterial catheter (Pulsiocath; Pulsion Medical Systems, Munich, Germany) inserted via femoral artery were present upon enrollment. The PiCCO device uses pulse contour analysis according to a modified algorithm originally described by Wesseling et al. [15] to determine PCCO and is described in more detail elsewhere [9]. This algorithm enables continuous calculation of stroke volume (SV) by measuring the systolic portion of the aortic pressure waveform and dividing the area under the curve by the aortic compliance. Therefore, the PiCCO device needs to be calibrated by COTCP. Calibrations were regularly performed by an ICU physician at defined time points (0:00 AM, 8:00 AM or 4:00 PM) with the patient in a supine position during a time period without acute hemodynamic instability using three subsequent boluses of 15 mL of ice-cold saline injected into the central venous line as proposed by the manufacturer [9]. During measurement, neither treatment provoking hemodynamic changes nor change of ventilation variables was performed. The dosage of vasopressors was kept constant. Our institutional guideline suggests calibration every 8 hours or before any major change in therapy is initiated. Therefore, additional calibrations by the attending ICU physician were allowed at any time. All hemodynamic data, including PCCO, central venous pressure (CVP), mean arterial blood pressure (MAP), pulse pressure (PP) (systolic minus diastolic aortic pressure) and heart rate (HR) were recorded immediately before and after calibration by COTCP. Global end-diastolic volume index (GEDI) and systemic vascular resistance index (SVRI) were derived upon thermodilution. SV was calculated as COTCP divided by heart rate. The PP to SV (PP/SV) relationship was used to examine the influence of NE dosage on central arterial stiffness as reported previously [16]. Our ICU is equipped with a patient data management system (PDMS) (CareSuite; Picis Inc., Wakefield, MA, USA) capable of electronically storing hemodynamic variables, including all single thermodilution calibrations, and ventilatory variables minute-by-minute.

Statistical analysis

Statistical analysis was performed using the statistical software R (R Foundation, Vienna, Austria [17]) and GraphPad Prism 5.01 software (GraphPad Software Inc., San Diego, CA, USA). Data are reported as means ± standard deviations (SD) unless otherwise specified. NE subgroups were defined as no NE, low-dose NE (<0.1 μg/kg/min) and high-dose NE (≥0.1 μg/kg/min) according to the Sepsis-Related Organ Failure Assessment score [18]. Subgroups of time interval elapsed after the latest calibration were defined as <2 hours, 2 to 4 hours, 4 to 8 hours, 8 to 16 hours and 16 to 24 hours. Data subsets for hemodynamic variables, PP/SV ratio and calibration interval were compared using an unpaired two-tailed t-test. Comparison of PCCO and COTCP was performed by using Bland-Altman statistics for multiple observations per individual [19], calculating mean differences between methods (bias) ±2 SD (limits of agreement). Bias between subgroups was compared using a t-test. The percentage error was calculated as reported by Critchley and Critchley [20], and interchangeability between methods was assumed as a percentage error below 30%. The precision of the reference technique (COTCP) was analyzed according to the method described by Cecconi et al. [21] from the three consecutive bolus injections for calibration. To test whether PCCO reflected changes (Δ) in CO, the ΔPCCO (PCCO - preceding COTCP) was analyzed against ΔCOTCP (actual COTCP - preceding COTCP) by linear regression analysis including the first pair of measurements of each patient. The influence of NE dosage and the severity of the patient's medical condition (APACHE II score) on calibration frequency was analyzed using the Spearman correlation for nonparametric data. P < 0.05 was considered statistically significant.

Results

Seventy-three patients were included in this study. The median (interquartile range) APACHE II score of all patients was 24 (range, 20 to 29) at the time of inclusion. Detailed patient characteristics are given in Table 1.

Table 1 Patient characteristics, medical history and reason for instrumentation with PiCCO monitoring systema

We obtained 330 data pairs. In 265 of 330 data pairs, patients received mechanical ventilation with a mean tidal volume of 8 ± 1 mL/kg, a mean fraction of inspired oxygen of 0.6 ± 0.1, a mean peak airway pressure of 23 ± 6 cmH2O and a mean positive end-expiratory pressure of 9 ± 3 cmH2O. In the remaining 65 data pairs, patients breathed spontaneously and received oxygen via face mask. Calibration interval was 9 ± 6 hours (range, 1 to 24 hours). The precision of the three bolus injection -COTCP values was 7%, according to the method of Cecconi et al. [21].

Concerning the effect of NE dosage on the agreement between PCCO and COTCP, 27 data pairs were excluded from further analysis because of additional dobutamine or epinephrine administration. In 161 of 303 data pairs, NE was administered in doses ranging from 0.01 to 4.29 μg/kg/min. The hemodynamic data and calibration intervals of different NE subgroups are presented in Table 2.

Table 2 Hemodynamic data and calibration interval of different norepinephrine subgroupsa

Bias between NE subgroups did not differ significantly. However, PCCO was interchangeable with COTCP only during high NE dosage and not at low or no NE dosage. The results of the Bland-Altman analysis are presented in Table 3, and plots are given in Figure 1.

Table 3 Results of Bland-Altman analysis of PCCO vs. COTCPa
Figure 1
figure 1

Bland-Altman plots of different norepinephrine (NE) subgroups. PCCO, pulse contour cardiac output; COTCP, transcardiopulmonary thermodilution cardiac output; PE, percentage error; solid line, mean bias; dotted lines, limits of agreement.

The coefficient of correlation values, r (95% confidence interval (95% CI)), between ΔPCCO and ΔCOTCP was 0.46 (95% CI, 0.25 to 0.64; P < 0.001) for all patients, 0.19 (95% CI, -0.23 to 0.55; P = 0.36) for no NE, 0.37 (95% CI, -0.09 to 0.70; P = 0.11) for NE < 0.1 μg/kg/min and 0.78 (95% CI, 0.53 to 0.91; P < 0.001) for NE ≥ 0.1 μg/kg/min subgroups, respectively. In the NE ≥ 0.1 μg/kg/min subgroup, a statistically significant (P < 0.05) higher PP/SV relationship (arterial stiffness) was observed compared to the no NE or NE < 0.1 μg/kg/min subgroups, respectively (Figure 2).

Figure 2
figure 2

Arterial stiffness. Pulse pressure (PP) to stroke volume (SV) relationship (PP/SV) as a measure of central arterial stiffness within the different norepinephrine (NE) dosage (μg/kg/min) subsets. Data are means ± SD; *P < 0.05 vs. no NE; #P < 0.05 vs. NE < 0.1 μg/kg/min.

The mean bias between PCCO and COTCP did not depend on time elapsed from the preceding calibration. However, in none of the subgroups did agreement between PCCO and COTCP meet defined criteria for interchangeability, as the percentage error was above 30% in all respective interval subgroups. The time-related effect on agreement is presented in Table 3. Individual bias during each interval, as well as mean bias ± limits of agreement, is plotted in Figure 3.

Figure 3
figure 3

Bias in relation to time interval between calibrations. Mean bias (boxes) ± limits of agreement and individual bias (circles) expressed as percentage of COTCP between PCCO and COTCP in subsets of different calibration intervals. Dotted lines illustrate interchangeability (±30%).

On our ICU, we recorded a mean (±SD) time interval after the preceding calibration of 9 ± 6 hours. In 151 (46%) recordings, the time interval exceeded the recommended 8-hour interval. In 14 (4%) recordings, the time interval was as long as 24 hours. The time interval did not correlate with NE dosage or APACHE II score (r = -0.04, P = 0.48; and r = -0.01, P = 0.41), respectively.

Discussion

In the present study, we have demonstrated an influence of NE dosage on agreement of PCCO, as only during high NE dosage the criteria of interchangeability with COTCP were met. Time elapsed between calibrations did not affect agreement between methods.

Goal-directed therapy in high-risk patients has been shown to improve outcomes [4, 5]. One essential observation in these studies was that the earlier treatment was started, the better the outcome. Therefore, continuous CO monitoring in critically ill patients is needed. However, PCCO needs to be validated in a large number of patients and during relevant conditions to gain more insight into the mechanisms influencing this variable. The present study compared PCCO and COTCP in 73 ICU patients with several comorbidities. Most previous studies compared PCCO with COTCP in small series of patients during cardiac surgery [6, 8, 9, 22]. Data from larger patient samples, however, are scarce. The percentage error between PCCO and CO derived by a thermodilution method varied between 26% and 50% in earlier studies [14, 23]. Critchley and Critchley [20] defined a percentage error of less than 30% to indicate interchangeability. Accordingly, we found an acceptable agreement of PCCO with COTCP only in data subsets obtained with high NE dosage, although a percentage error of 28% is still reasonably high. However, the results of the present study tend to refute our first hypothesis. Increasing NE dosage does not seem to be associated with decreased agreement between PCCO and COTCP, but rather with improved interchangeability. PCCO further showed a better performance in tracking changes in CO during increased NE dosage because the coefficient of correlation between ΔPCCO and ΔCOTCP was higher. Vascular tone seems to be an important issue regarding the agreement of PCCO methods with a reference method such as transcardiopulmonary thermodilution. Rodig et al. [12] described an increased bias between PCCO and CO measured by thermodilution after administration of phenylephrine. The observed change of SVR >60% between calibrations may explain their findings. A recent publication applying the same PCCO software used in our study concluded that agreement was not influenced by changes in SVR due to better adaptation of the newer algorithm [14]. In the present study, SVR was not different between NE subgroups. Therefore, we hypothesize that despite a comparable SVR, a differing compliance of the vascular tree between subgroups of different NE dosages may explain the different level of agreement. A higher NE dosage may result in an increased central arterial stiffness and therefore reduced arterial compliance [24], as recently reported by Wittrock et al. [16]. In agreement with these findings, high NE dosage resulted in a significantly higher PP/SV relationship as an indicator of arterial stiffness. Increasing arterial stiffness leads to a more rigid vascular system and therefore may result in better agreement between methods. It is conceivable in this context that the vasculature of patients on high NE has less oscillatory capacity, which limits changes in arterial compliance and consequently on the deviation from the compliance obtained upon calibration. In clinical practice, however, many patients may be treated with either a low dose of NE or no NE, and according to our results, PCCO is not interchangeable with COTCP in these patients.

Our results do not show a time-related effect on the agreement between PCCO and COTCP, thus refuting the second hypothesis. The percentage error was above 30% in all calibration interval subgroups. The manufacturer recommends recalibration every 8 hours. Godje et al. [9] reported an overall acceptable agreement up to 44 hours; however, they did not indicate the bias and percentage error of subsets regarding different calibration intervals. Hamzaoui et al. [14] reported a percentage error below 30% only within the first hour after calibration of PCCO, but up to 37% within a 6-hour calibration interval. These authors concluded that PCCO is stable during a 1-hour period, and even changes in SVR did not alter the agreement. These results would prompt one to use hourly recalibration. Regarding our results, time elapsed from preceding calibration did not determine the level of agreement, as individually good agreement was observed up to 24 hours and individually poor agreement occurred within a period of 2 hours after calibration. Moreover, we found acceptable agreement in patients who were administered a high NE dosage, and thus had higher arterial stiffness, who had mean calibration periods of 7 hours.

This study also examined the clinical use of calibrations by using PiCCO technology. Our institutional guidelines recommend a recalibration of the PiCCO system every 8 hours (three times daily), as well as before and after any major change in therapy. We found that in only 54% of recordings were institutional guidelines of recalibration met. We did not observe a correlation of calibration frequency with APACHE II score or NE dosage, indicating that calibration of PCCO may not be dependent on the severity of critical illness. These findings are surprising, since recalibration may increase agreement between methods [13]. However, our results indicate that the time interval between calibrations may not to be the most important factor in determining PCCO accuracy; moreover, therapy during calibrations seems to be important.

There are some limitations to our study. To avoid additional risk due to a more invasive methodology of CO measurement, we used the PiCCO integrated transcardiopulmonary thermodilution instead of the pulmonary artery thermodilution method as a reference technique for PCCO as previously described [13, 14]. The calibration interval was not strictly standardized to measure the effect of NE dosage on calibration frequency on our ICU.

Conclusions

This study demonstrates further limitations of the PCCO method for the determination of continuous CO. Only during high NE dosage (≥0.1 μg/kg/min) was PCCO interchangeable with COTCP. Therefore, the accuracy of PCCO measurement relies on important clinical circumstances.

Key messages

  • During clinical conditions, PCCO and COTCP measurements cannot be used interchangeably in patients who are either not on vasopressor treatment or on a low dose of vasopressors.

  • Acceptable agreement between the methods was observed only during an increased dose of norepinephrine, representing the minority of measurements. Even then the limits of agreement were rather large.

  • The time interval between calibrations of PCCO does not improve the reliability of PCCO within a period of 24 hours.

Abbreviations

Δ:

delta, change in CO between actual and preceding calibration

APACHE II:

Acute Physiology and Chronic Health Evaluation II score

CI:

cardiac index

CO:

cardiac output

COTCP:

transcardiopulmonary thermodilution cardiac output

CVP:

central venous pressure

GEDI:

global end-diastolic volume index

HR:

heart rate

ICU:

intensive care unit

MAP:

mean arterial pressure

NE:

norepinephrine

PCCO:

pulse contour cardiac output

PE:

percentage error

PP/SV:

pulse pressure to stroke volume ratio

r :

coefficient of correlation

SD:

standard deviation

SV:

stroke volume

SVRI:

systemic vascular resistance index.

References

  1. Rivers E, Nguyen B, Havstad S, Ressler J, Muzzin A, Knoblich B, Peterson E, Tomlanovich M: Early goal-directed therapy in the treatment of severe sepsis and septic shock. N Engl J Med 2001, 345: 1368-1377. 10.1056/NEJMoa010307

    Article  PubMed  CAS  Google Scholar 

  2. Eisenberg PR, Jaffe AS, Schuster DP: Clinical evaluation compared to pulmonary artery catheterization in the hemodynamic assessment of critically ill patients. Crit Care Med 1984, 12: 549-553. 10.1097/00003246-198407000-00001

    Article  PubMed  CAS  Google Scholar 

  3. Kern JW, Shoemaker WC: Meta-analysis of hemodynamic optimization in high-risk patients. Crit Care Med 2002, 30: 1686-1692. 10.1097/00003246-200208000-00002

    Article  PubMed  Google Scholar 

  4. McKendry M, McGloin H, Saberi D, Caudwell L, Brady AR, Singer M: Randomised controlled trial assessing the impact of a nurse delivered, flow monitored protocol for optimisation of circulatory status after cardiac surgery. BMJ 2004, 329: 258. 10.1136/bmj.38156.767118.7C

    Article  PubMed  PubMed Central  Google Scholar 

  5. Pearse R, Dawson D, Fawcett J, Rhodes A, Grounds RM, Bennett ED: Early goal-directed therapy after major surgery reduces complications and duration of hospital stay: a randomised, controlled trial [ISRCTN38797445]. Crit Care 2005, 9: R687-R693. 10.1186/cc3887

    Article  PubMed  PubMed Central  Google Scholar 

  6. Buhre W, Weyland A, Kazmaier S, Hanekop GG, Baryalei MM, Sydow M, Sonntag H: Comparison of cardiac output assessed by pulse-contour analysis and thermodilution in patients undergoing minimally invasive direct coronary artery bypass grafting. J Cardiothorac Vasc Anesth 1999, 13: 437-440. 10.1016/S1053-0770(99)90216-1

    Article  PubMed  CAS  Google Scholar 

  7. Bein B, Worthmann F, Tonner PH, Paris A, Steinfath M, Hedderich J, Scholz J: Comparison of esophageal Doppler, pulse contour analysis, and real-time pulmonary artery thermodilution for the continuous measurement of cardiac output. J Cardiothorac Vasc Anesth 2004, 18: 185-189. 10.1053/j.jvca.2004.01.025

    Article  PubMed  Google Scholar 

  8. Felbinger TW, Reuter DA, Eltzschig HK, Moerstedt K, Goedje O, Goetz AE: Comparison of pulmonary arterial thermodilution and arterial pulse contour analysis: evaluation of a new algorithm. J Clin Anesth 2002, 14: 296-301. 10.1016/S0952-8180(02)00363-X

    Article  PubMed  Google Scholar 

  9. Godje O, Hoke K, Goetz AE, Felbinger TW, Reuter DA, Reichart B, Friedl R, Hannekum A, Pfeiffer UJ: Reliability of a new algorithm for continuous cardiac output determination by pulse-contour analysis during hemodynamic instability. Crit Care Med 2002, 30: 52-58. 10.1097/00003246-200201000-00008

    Article  PubMed  Google Scholar 

  10. Felbinger TW, Reuter DA, Eltzschig HK, Bayerlein J, Goetz AE: Cardiac index measurements during rapid preload changes: a comparison of pulmonary artery thermodilution with arterial pulse contour analysis. J Clin Anesth 2005, 17: 241-248. 10.1016/j.jclinane.2004.06.013

    Article  PubMed  Google Scholar 

  11. Bein B, Meybohm P, Cavus E, Renner J, Tonner PH, Steinfath M, Scholz J, Doerges V: The reliability of pulse contour-derived cardiac output during hemorrhage and after vasopressor administration. Anesth Analg 2007, 105: 107-113. 10.1213/01.ane.0000268140.02147.ed

    Article  PubMed  Google Scholar 

  12. Rodig G, Prasser C, Keyl C, Liebold A, Hobbhahn J: Continuous cardiac output measurement: pulse contour analysis vs thermodilution technique in cardiac surgical patients. Br J Anaesth 1999, 82: 525-530.

    Article  PubMed  CAS  Google Scholar 

  13. Gruenewald M, Renner J, Meybohm P, Hocker J, Scholz J, Bein B: Reliability of continuous cardiac output measurement during intra-abdominal hypertension relies on repeated calibrations: an experimental animal study. Crit Care 2008, 12: R132. 10.1186/cc7102

    Article  PubMed  PubMed Central  Google Scholar 

  14. Hamzaoui O, Monnet X, Richard C, Osman D, Chemla D, Teboul JL: Effects of changes in vascular tone on the agreement between pulse contour and transpulmonary thermodilution cardiac output measurements within an up to 6-hour calibration-free period. Crit Care Med 2008, 36: 434-440. 10.1097/01.CCM.OB013E318161FEC4

    Article  PubMed  Google Scholar 

  15. Wesseling KH, Jansen JR, Settels JJ, Schreuder JJ: Computation of aortic flow from pressure in humans using a nonlinear, three-element model. J Appl Physiol 1993, 74: 2566-2573.

    PubMed  CAS  Google Scholar 

  16. Wittrock M, Scholze A, Compton F, Schaefer JH, Zidek W, Tepel M: Noninvasive pulse wave analysis for the determination of central artery stiffness. Microvasc Res 2009, 77: 109-112. 10.1016/j.mvr.2008.10.001

    Article  PubMed  Google Scholar 

  17. R Development Core Team: R: A language and environment for statistical computing. Vienna, Austria: R Foundation for Statistical Computing; 2009.

    Google Scholar 

  18. Vincent JL, Moreno R, Takala J, Willatts S, De Mendonça A, Bruining H, Reinhart CK, Suter PM, Thijs LG: The SOFA (Sepsis-related Organ Failure Assessment) score to describe organ dysfunction/failure. On behalf of the Working Group on Sepsis-Related Problems of the European Society of Intensive Care Medicine. Intensive Care Med 1996, 22: 707-710. 10.1007/BF01709751

    Article  PubMed  CAS  Google Scholar 

  19. Bland JM, Altman DG: Agreement between methods of measurement with multiple observations per individual. J Biopharm Stat 2007, 17: 571-582. 10.1080/10543400701329422

    Article  PubMed  Google Scholar 

  20. Critchley LA, Critchley JA: A meta-analysis of studies using bias and precision statistics to compare cardiac output measurement techniques. J Clin Monit Comput 1999, 15: 85-91. 10.1023/A:1009982611386

    Article  PubMed  CAS  Google Scholar 

  21. Cecconi M, Rhodes A, Poloniecki J, Della Rocca G, Grounds RM: Bench-to-bedside review: the importance of the precision of the reference technique in method comparison studies-with specific reference to the measurement of cardiac output. Crit Care 2009, 13: 201. 10.1186/cc7129

    Article  PubMed  PubMed Central  Google Scholar 

  22. Sander M, von Heymann C, Foer A, von Dossow V, Grosse J, Dushe S, Konertz WF, Spies CD: Pulse contour analysis after normothermic cardiopulmonary bypass in cardiac surgery patients. Crit Care 2005, 9: R729-R734. 10.1186/cc3903

    Article  PubMed  PubMed Central  Google Scholar 

  23. Ostergaard M, Nielsen J, Rasmussen JP, Berthelsen PG: Cardiac output-pulse contour analysis vs. pulmonary artery thermodilution. Acta Anaesthesiol Scand 2006, 50: 1044-1049. 10.1111/j.1399-6576.2006.01080.x

    Article  PubMed  CAS  Google Scholar 

  24. Chemla D, Hebert JL, Coirault C, Zamani K, Suard I, Colin P, Lecarpentier Y: Total arterial compliance estimated by stroke volume-to-aortic pulse pressure ratio in humans. Am J Physiol 1998, 274: H500-H505.

    PubMed  CAS  Google Scholar 

Download references

Acknowledgements

The authors thank Katja Frahm (physician), Sebastian Rossee and Moritz Maracke (both medical students) for excellent technical assistance. Funding was restricted to institutional and departmental sources. This work was presented in part at the American Society of Anesthesiologists Annual Meeting, October 2008, Orlando, FL, USA.

Author information

Authors and Affiliations

Authors

Corresponding author

Correspondence to Matthias Gruenewald.

Additional information

Competing interests

BB is a member of the advisory board of Pulsion Medical Systems. MG, PM, JR, AC, OB, NW, JS and MS declare that they have no competing interests.

Authors' contributions

MG conceived of the study design, carried out statistical analysis and drafted the manuscript. PM, OB and JR helped to draft the manuscript. AC supported statistical analysis. NW, JS and MS coordinated the study. BB conceived of the study design, coordinated the study and helped with statistical analysis and drafting of the manuscript. All authors read and approved the final manuscript.

Authors’ original submitted files for images

Below are the links to the authors’ original submitted files for images.

Authors’ original file for figure 1

Authors’ original file for figure 2

Authors’ original file for figure 3

Rights and permissions

Reprints and permissions

About this article

Cite this article

Gruenewald, M., Meybohm, P., Renner, J. et al. Effect of norepinephrine dosage and calibration frequency on accuracy of pulse contour-derived cardiac output. Crit Care 15, R22 (2011). https://doi.org/10.1186/cc9967

Download citation

  • Received:

  • Revised:

  • Accepted:

  • Published:

  • DOI: https://doi.org/10.1186/cc9967

Keywords