Table 3 Results of Bland-Altman analysis of PCCO vs. COTCPa

From: Effect of norepinephrine dosage and calibration frequency on accuracy of pulse contour-derived cardiac output

 

Number of patients

Mean

Bias

Limits of agreement

Percentage error

Parameter

( n all / n patient )

(L/min)

(L/min)

(L/min)

(%)

All

330/73

8.1

0.16

-2.81-3.15

38

No NE

142/44

8.41

0.16

-3.12-3.44

40

NE < 0.1 (μg/kg/min)

82/38

8.50

0.06

-3.88-4.00

47

NE ≥ 0.1 (μg/kg/min)

79/30

7.87

0.29

-1.83-2.42

28b

Calibration interval 0 to 2 hours

36/25

8.00

0.25

-4.00-4.51

54

Calibration interval 2 to 4 hours

48/35

7.78

0.12

-3.37-3.60

46

Calibration interval 4 to 8 hours

95/41

8.21

0.09

-2.43-2.61

31

Calibration interval 8 to 16 hours

101/47

8.19

0.21

-3.17-3.59

42

Calibration interval 16 to 24 hours

50/28

8.06

0.23

-2.90-3.34

40

  1. anall, number of measurement pairs for pulse contour-derived cardiac output (PCCO) and transcardiopulmonary thermodilution cardiac output (COTCP); npatient, number of patients; mean, mean of all PCCO and COTCP measurements. bInterchangeability according to Critchley and Critchley [20]. Bias and limits of agreement were calculated according to the method of Bland and Altman [19].