Nitrogen washout/washin, helium dilution and computed tomography in the assessment of end expiratory lung volume
Critical Care volume 12, Article number: R150 (2008)
Abstract
Introduction
End expiratory lung volume (EELV) measurement in the clinical setting is routinely performed using the helium dilution technique. A ventilator that implements a simplified version of the nitrogen washout/washin technique is now available. We compared the EELV measured by spiral computed tomography (CT) taken as gold standard with the lung volume measured with the modified nitrogen washout/washin and with the helium dilution technique.
Methods
Patients admitted to the general intensive care unit of Ospedale Maggiore Policlinico Mangiagalli Regina Elena requiring ventilatory support and, for clinical reasons, thoracic CT scanning were enrolled in this study. We performed two EELV measurements with the modified nitrogen washout/washin technique (increasing and decreasing inspired oxygen fraction (FiO2) by 10%), one EELV measurement with the helium dilution technique and a CT scan. All measurements were taken at 5 cmH2O airway pressure. Each CT scan slice was manually delineated and gas volume was computed with custom-made software.
Results
Thirty patients were enrolled (age = 66 +/- 10 years, body mass index = 26 +/- 18 Kg/m2, male/female ratio = 21/9, partial arterial pressure of carbon dioxide (PaO2)/FiO2 = 190 +/- 71). The EELV measured with the modified nitrogen washout/washin technique showed a very good correlation (r2 = 0.89) with the data computed from the CT with a bias of 94 +/- 143 ml (15 +/- 18%, p = 0.001), within the limits of accuracy declared by the manufacturer (20%). The bias was shown to be highly reproducible, either decreasing or increasing the FiO2 being 117+/-170 and 70+/-160 ml (p = 0.27), respectively. The EELV measured with the helium dilution method showed a good correlation with the CT scan data (r2 = 0.91) with a negative bias of 136 +/- 133 ml, and appeared to be more correct at low lung volumes.
Conclusions
The EELV measurement with the helium dilution technique (at low volumes) and modified nitrogen washout/washin technique (at all lung volumes) correlates well with CT scanning and may be easily used in clinical practice.
Trial Registration
Current Controlled Trials NCT00405002.
Introduction
The damage induced by mechanical ventilation in cases of acute lung injury (ALI) or acute respiratory distress syndrome (ARDS) can be termed barotrauma [1], volotrauma [2, 3], atelectrauma [4, 5] or biotrauma [5, 6] depending on the emphasis given to the pathogenic mechanism. They are all caused by the unphysiological stress and strain applied to the whole lung or to particular regions of the ventilated lung [7]. Stress and strain are linked by the specific lung elastance, which has a constant proportionality function [8]. Consequently, knowing the lung stress (i.e. the transpulmonary pressure) or the strain (i.e. the change in volume of the lung relative to its rest position, the functional residual capacity), allows us to better define the mechanical characteristics of the system and to design a safer mechanical ventilation. Although the stress requires the measurement of the oesophageal pressure, the strain requires the measurement of the lung volume.
In the intensive care unit (ICU) setting the measurement of lung volume is not routinely performed; in our practice Ospedale Maggiore Policlinico Mangiagalli Regina Elena the helium dilution technique, which requires an appreciable amount of time and work to become accomplished in, has been used for several years [9, 10]. A new technique has been recently proposed (LUFU, acronym for LUng FUnction) based on a modified nitrogen washout/washin technique using a side stream fast oxygen analyser [11]. It is not yet commercially available, but shows a good accuracy in lung volume measurement when compared with the helium dilution technique [12, 13].
Another modification of the nitrogen washout/washin technique has been implemented in mechanical ventilation (Engstrom Carestation) [14]. This technique is the subject of the present investigation. The ventilator in which the technique is implemented is equipped with a supplemental pressure port through which the oesophageal pressure could be continuously measured breath by breath. Indeed this ventilator incorporates the technology for measuring both the lung stress (the transpulmonary pressure) and the lung strain (volume change divided by the functional residual capacity) [8, 15]. Therefore, to assess the accuracy of the Engstrom Carestation to measure gas volume (and strain) we set up a comparative study. The end expiratory lung volume (EELV), at 5 cmH2O airway pressure, was measured by the helium dilution technique, nitrogen washout/washin technique and by computed tomography (CT) scanning, which was taken as the reference gold standard [16].
Materials and methods
Study population
Measurements were taken from 30 patients admitted to the ICU from November 2006 to September 2007. The study was approved by the institutional review board of our hospital, and written informed consent was obtained from conscious patients and delayed consent in unconscious patients. Inclusion criteria was the clinical need of a lung CT scan in patients already in mechanical ventilation. Exclusion criteria were age younger than 16 years, pregnancy, haemodynamic instability, documented barotrauma and the presence of chronic lung disease (e.g. chronic obstructive pulmonary disease (COPD)).
Data collection
The CT scan was performed at end expiration at 5 cmH2O positive end expiratory pressure (PEEP), which is the standard pressure at which CT scans are taken for clinical purposes. Thereafter, the patients underwent the measurement of the EELV with the nitrogen washout/washin technique performed with the Engstrom Carestation ventilator and with helium dilution at 5 cmH2O PEEP. The nitrogen washout/washin technique was performed by either decreasing or increasing the inspired oxygen fraction (FiO2). The average result of these two measurements allowed the accuracy of the Engstrom Carestation in relation to the CT scan to be assessed. Each sequence (decreasing or increasing the FiO2) was performed twice to assess the precision (repeatability) of the technique. The accuracy of the helium dilution technique was assessed relative to the CT scan measurement. Only one CT measurement was performed, as the precision of this method has previously been determined [17].
Quantitative computed tomography analysis
The CT scanner was set as follows: collimation 5 mm; interval 5 mm; bed speed 15 mm per second; voltage 140 kV; and current 240 mA. A whole lung CT scan was performed at a PEEP value of 5 cmH2O during an end-expiratory pause. Immediately before each CT scan was obtained, a recruitment manoeuvre was performed. Lungs profile was manually delineated in each cross-sectional lung image in order to identify the regions of interest. Each region of interest was then processed and analysed by a custom-designed software package (Soft-E-Film, University of Milan, Italy), as previously described [18].
We assumed that lung tissue has a density similar to that of water (Hounsefield unit number 0) and considered each voxel to be made of lung tissue and air (Hounsefield unit number -1000). Gas volume can be computed from the Hounsefield number of each voxel according to the following formula:
Gas volume = (CT number/-1000) × voxel volume
The EELV is the sum of gas volumes present in all the voxel included in the lung profiles.
Nitrogen washout/washin technique
The nitrogen washout/washin technique is based on the following principle: the gas lung volume, at baseline, includes a volume of nitrogen (V(1)N2) that is determined by the alveolar fraction of nitrogen (FAN2(1)) (which varies inversely to the alveolar oxygen fraction) and by the EELV accordingly to the following relation:
V(1)N2 = FAN2(1) × EELV
If the alveolar nitrogen fraction (FAN2(2)) is changed by changing the FiO2, a new nitrogen volume (V(2)N2) will be present in the lung after the equilibrium time:
V(2)N2 = FAN2(2) × EELV
Assuming that after changing the FiO2 the total EELV does not change until the new equilibrium in alveolar gas composition is reached, by subtracting term by term in the equation 1 and 2 the following relation holds true:
VN2(1) - VN2(2) = (FAN2(2) - FAN2(1)) × EELV
as the changes in FAN2 are specular to the changes in FiO2, i.e. ΔFAN2 = -(FiO2(1) - FiO2(2)), the EELV can be calculated as:
EELV = ΔN2 (ml)/ΔFiO2
where ΔN2 equals the nitrogen exhaled after the change of inspired FiO2 until the equilibration time is reached (about 20 breaths).
The algorithm of the nitrogen washout/washin technique employed by the Engstrom Carestation is detailed by Olegard and colleagues [14]. Nitrogen concentration in expired and inspired air is not directly measured but estimated from the end tidal concentrations of oxygen and carbon dioxide:
ETN2 (mmHg) = 713 - ETCO2 (mmHg) - ETO2 (mmHg)
The alveolar ventilation was calculated as:
Alveolar tidal volume expired = VCO2/ETCO2 × RR
Alveolar tidal volume inspired = Alveolar tidal volume inspired + ((VCO2/RQ + VCO2)/RR)
Inspired and expired nitrogen volumes were calculated as:
Expired nitrogen volume = ETN2/713 × Alveolar tidal volume expired
Inspired tidal volume = Inspired nitrogen fraction × alveolar tidal volume inspired.
Simplified helium dilution technique
A flexible tube was inserted between the Y-piece and the patient's endotracheal tube or tracheostomy. The operator clamped the tube during an end-expiratory pause at a PEEP level of 5 cmH2O and then connected it to a balloon filled with 1.5 L of a gas mixture of helium (13.07 ± 0.40%) in oxygen. After releasing the clamp, the same operator delivered 10 tidal volumes to the patient in order to dilute the helium gas mixture with the gas contained in the patient's lungs. At the end of this procedure, the balloon was clamped off the circuit, and the patient was reconnected to the ventilator. The concentration of helium in the balloon was then measured by a previously calibrated helium analyser (PK Morgan, Chatham, England). EELV was then calculated using the standard formula: EELV (ml) = (Vb × Ci/Cf) - Vb, where Ci is the helium concentration of the known gas mixture, Cf is the final helium concentration and Vc is the volume of gas in the balloon. The Vb was inflated with 1500 ml of helium-oxygen mixture at 25°C. The volume measured was corrected for body temperature (37°C) using the Gay-Lussac law.
We first performed the CT scan, then the helium dilution technique and at the end the modified nitrogen washout/washin technique. The helium dilution technique and the modified nitrogen washout/washin technique were performed at the CT scan facility without moving the patient from the bed, to maintain the patient's condition. The whole experimental procedure (nitrogen washout/washin technique, helium dilution and CT scan) was performed in a total time of about five minutes.
Lung mechanics
The total inspiratory resistance of the respiratory system was calculated by dividing the difference in peak inspiratory airway pressure and the plateau inspiratory pressure, measured during an end inspiratory pause, by the inspiratory flow preceding the occlusion. The compliance of the respiratory system was calculated by dividing the plateau inspiratory pressure measured during an end inspiratory pause by the tidal volume.
Statistical methods
Gas volumes measured with CT scan, nitrogen washout/washin technique and helium dilution technique were compared with the Bland-Altman technique [19] and using a linear regression model. The bias of the EELV measurement performed increasing or decreasing the FiO2 were compared using a Student's t-test. Statistical analysis was performed with the R-project software (R foundation for statistical computing, Vienna, Austria [20]).
Results
Table 1 summarises the clinical characteristics of the patient population. As shown, 66% of the patients could be classified as having ALI with ARDS, 23% could be classified as having ALI without ARDS and 10% had a PaO2/FiO2 ratio greater than 300.
CT scan and nitrogen washout/washin
As shown in Figure 1a, the EELV measured by the Engstrom Carestation (as the average of two measurements at different FiO2), was highly correlated with the EELV computed by the CT scan (r2 = 0.89). The Bland-Altman plot (Figure 1b) revealed a bias of 94 ± 143 ml (15 ± 18%, p < 0.001). The relative error of the simplified nitrogen washout/washin technique (expressed as (EELVGE - EELVCT SCAN)/EELVCT SCAN) was significantly related to the ratio between TV/EELVCT SCAN (y = (0.05 + x) × 0.43, r2 = 0.58), as shown in Figure 2.
The accuracy of the nitrogen washout/washin method was similar when either increasing or decreasing the FiO2, with a bias of 117 ± 170 and 70 ± 160 (p = 0.27), respectively. The precision of the nitrogen washout/washin measurements is shown in Figure 3 which underlines the high reproducibility of the method with a difference between the two measurements of 48 ± 165 ml, which was not statistically different from zero (p = 0.12).
CT scan and helium dilution
As shown in Figure 4a, the EELV measured by the helium dilution technique was highly correlated (r2 = 0.91) with the EELV computed by the CT scan. The Bland-Altman plot (Figure 4b) revealed a negative bias of -136 ± 133 ml (16 ± 13%, p < 0.001). The Bland-Altman plot also showed a significant negative correlation between the increase of EELV and the difference between the volume measured by the CT scan and the volume measured by the helium dilution technique. Indeed the EELV measured by the helium dilution technique is more accurate at low lung volumes and the accuracy of the measurement decreases when the gas lung volume increases.
Nitrogen washout/washin and helium dilution techniques
As shown in Figure 5a, the EELV measured by helium dilution technique was well correlated (r2 = 82) with the EELV calculated with the nitrogen washout/washin technique with a negative bias of -229 ± 164 ml (40 ± 26%, p < 0.001; Figure 5b).
Discussion
The primary finding of this study is that the three different methods we tested (CT scan, simplified nitrogen washout/washin technique and helium dilution technique) to measure the EELV in critically ill patients are in reasonable agreement.
The CT scan measures the lung density and estimates gas volume assuming that the lung is composed of two compartments with very different densities: lung "tissue" and gas [18]. Consequently, CT scans accurately estimate the lung inflation independent of how the different lung regions are ventilated. In contrast, both the nitrogen washout/washin and helium dilution techniques for EELV determination rely on ventilation and measure the fraction of EELV that is ventilable; that is, non-ventilated or poorly ventilated lung compartments will be excluded or underestimated. In this study this fraction is likely to be of minor importance, because we tried to exclude patients with a diagnosis of COPD or other airway disease, in whom the difference between CT measurement and the ventilation-based techniques may be relevant, from our study.
Nitrogen dilution technique was first employed by Durig in 1903 [21] but nitrogen washout/washin techniques did not gain widespread application outside the research setting because nitrogen must be measured with a mass spectrometer. Fretschner and colleagues in 1993 proposed a method to measure the EELV without the need to directly measure the nitrogen concentration [22], but it relied on oxygen and carbon dioxide measurement. The proposed method, however, implied a 30% change in FiO2 and the synchronisation between gas measurement and flow measurement. The methodology described by Olegard and colleagues [14] and implemented by the Engstrom Carestation overcomes these two limitations. Synchronisation problems are avoided by calculating the alveolar nitrogen concentration from end-tidal carbon dioxide and oxygen concentrations, and the FiO2 changes are limited to only 10%, which can be considered safe even in critically ill patients. In our experimental setting this simplified nitrogen dilution technique showed an average overestimation of lung volume of 94 ± 143 ml (15 ± 18%, p < 0.001), within the limits of accuracy declared by the manufacturer (20%). The technique also proved to be highly reproducible, either by increasing or decreasing the FiO2. Indeed it appears to be suitable for clinical application.
In this study we found that the simplified helium dilution technique had a negative bias of 136 ± 133 ml (16 ± 13%) in respect to CT scan; the bias of the helium dilution technique increased linearly with increasing EELV. The simplified helium dilution technique we used, however, was developed to measure the small lung volume of ARDS patients ('baby lung'); it is possible that in patients with a higher EELV the equilibrium in the system (balloon plus lungs) is not reached with 10 normal breaths. The equilibrium time is the function of the time constant of the system (i.e. the time taken to reach approximately 63% of its final (asymptotic) value), which is the product of the compliance of the respiratory system and airway resistances. The time constant is short in ALI/ARDS patients (low compliance), although it is high in patients with COPD, emphysema or any condition that increases lung compliance and airway resistance.
The helium dilution technique differs from the nitrogen washout/washin technique in that no fresh gas ventilation occurs and a reduction in lung volume due to oxygen absortive phenomena is possible, even if data present in the literature show that its effect should be negligible [23, 24]. Accordingly the difference between the EELV measured with the CT scan and the EELV measured with the helium dilution method was negligible at lower lung volumes and become consistent increasing the EELV. The helium dilution technique was performed in 10 breaths, while the simplified nitrogen washout/washin technique required 20 breaths to reach equilibrium and this may in part account for the difference between the two techniques based on ventilation. Moreover, the helium dilution method is prone to other sources of error such as the helium concentration in the gas tank, which usually has a tolerance degree of about 5%, and possible diffusion of helium out of the balloon. In addition the possibility of helium uptake during EELV measurement has been discussed in the literature and is thought to be controversial [25–27]. Theoretical considerations show that the possible helium uptake is limited: the solubility coefficient of helium is 1.13 × 10-5 ml/ml/mmHg [28] with a maximal body helium uptake of about 45 ml in 40 L body water and about 5 ml in well perfused body compartments.
Conclusion
The CT remains the gold standard for measuring gas lung volume. The helium dilution technique is clinically acceptable when applied in patients with a short time constant of the respiratory system. The nitrogen dilution technique appears to be simple and effective.
Key messages
-
The EELV measured with the modified nitrogen washout/washin technique well correlated with the CT scan and is useful for clinical purposes.
-
The EELV with helium dilution technique correlated well with the CT scan at low lung volumes but showed a systematic underestimation at greater lung volumes.
Abbreviations
- ALI:
-
acute lung injury
- ARDS:
-
acute respiratory distress syndrome
- COPD:
-
chronic obstructive pulmonary disease
- CT:
-
computed tomography
- EELV:
-
end expiratory lung volume
- FiO2 :
-
inspired oxygen fraction
- PEEP:
-
positive end expiratory pressure
- ICU:
-
intensive care unit.
References
Kolobow T, Moretti MP, Fumagalli R, Mascheroni D, Prato P, Chen V, Joris M: Severe impairment in lung function induced by high peak airway pressure during mechanical ventilation. An experimental study. Am Rev Respir Dis 1987, 135: 312-315.
Dreyfuss D, Basset G, Soler P, Saumon G: Intermittent positive-pressure hyperventilation with high inflation pressures produces pulmonary microvascular injury in rats. Am Rev Respir Dis 1985, 132: 880-884.
Dreyfuss D, Soler P, Basset G, Saumon G: High inflation pressure pulmonary edema. Respective effects of high airway pressure, high tidal volume, and positive end-expiratory pressure. Am Rev Respir Dis 1988, 137: 1159-1164.
Mead J, Takishima T, Leith D: Stress distribution in lungs: a model of pulmonary elasticity. J Appl Physiol 1970, 28: 596-608.
Tremblay L, Valenza F, Ribeiro SP, Li J, Slutsky AS: Injurious ventilatory strategies increase cytokines and c-fos m-RNA expression in an isolated rat lung model. J Clin Invest 1997, 99: 944-952. 10.1172/JCI119259
Chiumello D, Pristine G, Slutsky AS: Mechanical ventilation affects local and systemic cytokines in an animal model of acute respiratory distress syndrome. Am J Respir Crit Care Med 1999, 160: 109-116.
Gattinoni L, Carlesso E, Cadringher P, Valenza F, Vagginelli F, Chiumello D: Physical and biological triggers of ventilator-induced lung injury and its prevention. Eur Respir J Suppl 2003, 47: 15s-25s. 10.1183/09031936.03.00021303
Chiumello D, Carlesso E, Cadringher P, Caironi P, Valenza F, Polli F, Tallarini F, Cozzi P, Cressoni M, Colombo A, Marini JJ, Gattinoni L: Lung stress and strain during mechanical ventilation for acute respiratory distress syndrome. Am J Respir Crit Care Med 2008, 178: 346-355. 10.1164/rccm.200710-1589OC
Gattinoni L: Metodo semplificato per la determinazione della capacità funzionale residua. Anestesia e Rianimazione 1979, 20: 103-114.
Damia G, Mascheroni D, Croci M, Tarenzi L: Perioperative changes in functional residual capacity in morbidly obese patients. Br J Anaesth 1988, 60: 574-578. 10.1093/bja/60.5.574
Mitchell RR, Wilson RM, Holzapfel L, Benis AM, Sierra D, Osborn JJ: Oxygen wash-in method for monitoring functional residual capacity. Crit Care Med 1982, 10: 529-533.
Maisch S, Boehm SH, Weismann D, Reissmann H, Beckmann M, Fuellekrug B, Meyer A, Schulte Am EJ: Determination of functional residual capacity by oxygen washin-washout: a validation study. Intensive Care Med 2007, 33: 912-916. 10.1007/s00134-007-0578-2
Patroniti N, Saini M, Zanella A, Weismann D, Isgro S, Bellani G, Foti G, Pesenti A: Measurement of end-expiratory lung volume by oxygen washin-washout in controlled and assisted mechanically ventilated patients. Intensive Care Med 2008, 34: 2235-2240. 10.1007/s00134-008-1218-1
Olegard C, Sondergaard S, Houltz E, Lundin S, Stenqvist O: Estimation of functional residual capacity at the bedside using standard monitoring equipment: a modified nitrogen washout/washin technique requiring a small change of the inspired oxygen fraction. Anesth Analg 2005, 101: 206-212. 10.1213/01.ANE.0000165823.90368.55
Gattinoni L, Carlesso E, Cadringher P, Valenza F, Vagginelli F, Chiumello D: Physical and biological triggers of ventilator-induced lung injury and its prevention. Eur Respir J Suppl 2003, 47: 15s-25s. 10.1183/09031936.03.00021303
Gattinoni L, Chiumello D, Cressoni M, Valenza F: Pulmonary computed tomography and adult respiratory distress syndrome. Swiss Med Wkly 2005, 135: 169-174.
Patroniti N, Bellani G, Manfio A, Maggioni E, Giuffrida A, Foti G, Pesenti A: Lung volume in mechanically ventilated patients: measurement by simplified helium dilution compared to quantitative CT scan. Intensive Care Med 2004, 30: 282-289. 10.1007/s00134-003-2109-0
Gattinoni L, Mascheroni D, Torresin A, Marcolin R, Fumagalli R, Vesconi S, Rossi GP, Rossi F, Baglioni S, Bassi F: Morphological response to positive end expiratory pressure in acute respiratory failure. Computerized tomography study. Intensive Care Med 1986, 12: 137-142. 10.1007/BF00254928
Bland JM, Altman DG: Statistical methods for assessing agreement between two methods of clinical measurement. Lancet 1986, 1: 307-310.
The R Project for Statistical Computing[http://www.R-project.org]
Durig A: Uber die Grosse der Residualluft. Zentralblatt Physiol 1903, 17: 258-267.
Fretschner R, Deusch H, Weitnauer A, Brunner JX: A simple method to estimate functional residual capacity in mechanically ventilated patients. Intensive Care Med 1993, 19: 372-376. 10.1007/BF01724875
Chiumello D, Carlesso E, Aliverti A, Dellaca RL, Pedotti A, Pelosi PP, Gattinoni L: Effects of volume shift on the pressure-volume curve of the respiratory system in ALI/ARDS patients. Minerva Anestesiol 2007, 73: 109-118.
Gattinoni L, Mascheroni D, Basilico E, Foti G, Pesenti A, Avalli L: Volume/pressure curve of total respiratory system in paralysed patients: artefacts and correction factors. Intensive Care Med 1987, 13: 19-25. 10.1007/BF00263552
Meneely GR, Kaltreider NL: The volume of the lung determined by helium dilution. Description of the method and comparison with other procedures. J Clin Invest 1949, 28: 129-139. 10.1172/JCI102041
Holmgren A: Determination of the functional residual volume by means of the helium dilution method. Scand J Clin Lab Invest 1954, 6: 131-136. 10.3109/00365515409134854
Birath G, Swenson EW: A correction factor for helium absorption in lung volume determinations. Scand J Clin Lab Invest 1956, 8: 155-158. 10.3109/00365515609049261
Hawkins JA, Shilling CW: Helium solubility in blood at increased pressures. J Biol Chem 1936, 113: 649-653.
Acknowledgements
We are thankful to all the staff of the "E. Vecla" ICU.
Author information
Authors and Affiliations
Corresponding author
Additional information
Competing interests
LG received lecture fees from GE Healthcare.
Authors' contributions
DC was responsible for patient selection, clinical conduction of the study and participated in writing the manuscript. MC prepared the database, analysed the data and drafted the manuscript. MC participated in clinical conduction of the study. FT participated in clinical conduction of the study. MB, VB and CM analysed the CT scan data. LG was responsible for study design and participated in writing the manuscript. All the authors read and approved the final manuscript.
An erratum to this article is available at http://dx.doi.org/10.1186/cc7743.
Authors’ original submitted files for images
Below are the links to the authors’ original submitted files for images.
Rights and permissions
This article is published under license to BioMed Central Ltd. This is an Open Access article distributed under the terms of the Creative Commons Attribution License (http://creativecommons.org/licenses/by/2.0/), which permits unrestricted use, distribution, and reproduction in any medium, provided the original work is properly cited.
About this article
Cite this article
Chiumello, D., Cressoni, M., Chierichetti, M. et al. Nitrogen washout/washin, helium dilution and computed tomography in the assessment of end expiratory lung volume. Crit Care 12, R150 (2008). https://doi.org/10.1186/cc7139
Received:
Revised:
Accepted:
Published:
DOI: https://doi.org/10.1186/cc7139