Figure 1From: Nitrogen washout/washin, helium dilution and computed tomography in the assessment of end expiratory lung volumeComparison of end expiratory lung volume (EELV) measured by the Engstrom Carestation and the computed tomography (CT) scan. (a)The EELV measured by the Engstrom Carestation as a function of the EELV measured by the computed tomography (CT) scan (EELV carestation = 242 + 0.85 × EELV CT scan, r2 = 0.89, p < 0.00001). (b) The Bland-Altman plot of the EELV measured with the CT scan and the EELV measured with the Engstrom Carestation. The x axis shows the mean of the two measurement and the the y axis shows the difference between the EELV measured by the Engstrom Carestation and the EELV measured by the CT scan (average difference 93 ± 143 ml, limits of agreement -50 – 236 ml).Back to article page