Skip to main content
  • Letter
  • Published:

Measuring extravascular lung water: animals and humans are not the same

The evolution of extravascular lung water (EVLW) monitoring is an important step forward in the hemodynamic assessment of critically ill patients.

The determination of EVLW with single transpulmonary thermodilution (STD) has shown an acceptable accuracy and reproducibility [14]. However, in the experimental setting the STD technique can overestimate EVLW compared with postmortem gravimetry as a reference method [13]. The calculation of EVLW is based on the assumption that EVLW = intrathoracic thermal volume (ITTV) – intrathoracic blood volume (ITBV). According to the STD algorithm (PiCCO; Pulsion Medical Systems, Munich, Germany), in humans ITBV = 1.25 × global end-diastolic volume (GEDV) – 28.4 (ml) [4]. However, in experimental studies the relationship between ITBV and GEDV differs significantly. It is still unsettled whether this depends on species differences or different body weights. In pigs weighing between 24.0 and 31.8 kg, the linear regression equation is ITBV = 1.52 × GEDV – 49.7 [1]. Replacing the default PiCCO algorithm by the latter equation, the overestimation was reduced to a level that was not significantly different from the gravimetric value.

Employing a thermal dye technique (TDD, Cold Z-021; Pulsion Medical Systems), we recently determined ITBV, GEDV, the ITBV:GEDV ratio, and pulmonary blood volume (PBV) in sheep. The measurements were performed during spontaneous breathing, during mechanical ventilation, and after pneumonectomy (Table 1). Our results confirm the close correlation between ITBV and GEDV in different states. They may also indicate a relationship between ventilation, perfusion and the ITBV/GEDV ratio [4]. Mechanical ventilation can decrease preloading, thus reducing GEDV, whereas ITBV continues to remain an accurate marker of preload. Most probably, ITBV reduces after pneumonectomy because of a decrease in PBV. We therefore suggest that the ITBV:GEDV ratio depends on relative and independent changes in ITBV and GEDV.

Table 1 Volumetric variables in sheep breathing spontaneously, or subjected to mechanical ventilation, or pneumonectomy

We agree with other investigators [1, 2] that STD may be a useful tool for monitoring changes in EVLW over time. Our recent study [5] demonstrates that in patients with severe sepsis EVLW measured by STD has a prognostic role and is significantly correlated with lung injury score. Although we have to keep in mind the technical aspects of STD that can differ between experimental and clinical settings, EVLW has evolved as a key variable for evaluating pulmonary hemodynamics. Thus, monitoring of EVLW might be a valid dynamic measure of lung edema at the bedside, supporting therapeutical decisions on patients with a compromised cardiopulmonary function.

Key messages

  • The calculation of extravascular lung water with the use of single transpulmonary thermodilution in the experimental animals requires a specific correction.

  • The monitoring of extravascular lung water is a valid dynamic method for measuring lung edema at the bedside.

Abbreviations

EVLW:

extravascular lung water

GEDV:

global end-diastolic volume

ITBV:

intrathoracic blood volume

PBV:

pulmonary blood volume

STD:

single transpulmonary thermodilution.

References

  1. Rossi P, Wanecek M, Rudehill A, Konrad D, Weitzberg E, Oldner A: Comparison of a single indicator and gravimetric technique for estimation of extravascular lung water in endotoxemic pigs. Crit Care Med 2006, 34: 1437-1443. 10.1097/01.CCM.0000215830.48977.29

    Article  PubMed  Google Scholar 

  2. Katzenelson R, Perel A, Berkenstadt H, Preisman S, Kogan S, Sternik L, Segal E: Accuracy of transpulmonary thermodilution versus gravimetric measurement of extravascular lung water. Crit Care Med 2004, 32: 1550-1554. 10.1097/01.CCM.0000130995.18334.8B

    Article  PubMed  Google Scholar 

  3. Kirov MY, Kuzkov VV, Kuklin VN, Waerhaug K, Bjertnaes LJ: Extravascular lung water assessed by transpulmonary single thermodilution and postmortem gravimetry in sheep. Crit Care 2004, 8: R451-R458. 10.1186/cc2974

    Article  PubMed Central  PubMed  Google Scholar 

  4. Sakka SG, Ruhl CC, Pfeiffer UJ, Beale R, McLuckie A, Reinhart K, Meier-Hellmann A: Assessment of cardiac preload and extravascular lung water by single transpulmonary thermodilution. Intensive Care Med 2000, 26: 180-187. 10.1007/s001340050043

    Article  CAS  PubMed  Google Scholar 

  5. Kuzkov VV, Kirov MY, Sovershaev MA, Kuklin VN, Suborov EV, Waerhaug K, Bjertnaes LJ: Extravascular lung water determined with single transpulmonary thermodilution correlates with the severity of sepsis-induced lung injury. Crit Care Med 2006, 34: 1647-1653. 10.1097/01.CCM.0000218817.24208.2E

    Article  PubMed  Google Scholar 

Download references

Acknowledgements

The studies in sheep described in this letter were approved by the Nor-wegian Experimental Animal Board and performed in compliance with the European Convention on Animal Care at the University of Tromsø, Norway.

Author information

Authors and Affiliations

Authors

Corresponding author

Correspondence to Mikhail Y Kirov.

Additional information

Competing interests

MYK and EF-M are members of the Medical Advisory Board, Pulsion Medical Systems. Approximately 25% of the total expenses to two of the studies upon which this letter is based has been reimbursed by Pulsion Medical Systems, Munich, Germany.

Authors' contributions

MYK and VVK participated in the design of the study, performed the statistical analysis, and drafted the manuscript. EF-M and LJB participated in the preparation of the draft. All authors read and approved the final version.

Rights and permissions