Effects of positive end-expiratory pressure on gastric mucosal perfusion in acute respiratory distress syndrome

Abstract

Introduction

Positive end-expiratory pressure (PEEP) improves oxygenation and can prevent ventilator-induced lung injury in patients with acute respiratory distress syndrome (ARDS). Nevertheless, PEEP can also induce detrimental effects by its influence on the cardiovascular system. The purpose of this study was to assess the effects of PEEP on gastric mucosal perfusion while applying a protective ventilatory strategy in patients with ARDS.

Methods

Eight patients were included. A pressure–volume curve was traced and ideal PEEP, defined as lower inflection point + 2 cmH2O, was determined. Gastric tonometry was measured continuously (Tonocap). After baseline measurements, 10, 15 and 20 cmH2O PEEP and ideal PEEP were applied for 30 min each. By the end of each period, hemodynamic, CO2 gap (gastric minus arterial partial pressures), and ventilatory measurements were performed.

Results

PEEP had no effect on CO2 gap (median [range], baseline: 19 [2–30] mmHg; PEEP 10: 19 [0–40] mmHg; PEEP 15: 18 [0–39] mmHg; PEEP 20: 17 [4–39] mmHg; ideal PEEP: 19 [9–39] mmHg; P = 0.18). Cardiac index also remained unchanged (baseline: 4.6 [2.5–6.3] l min-1 m-2; PEEP 10: 4.5 [2.5–6.9] l min-1 m-2; PEEP 15: 4.3 [2–6.8] l min-1 m-2; PEEP 20: 4.7 [2.4–6.2] l min-1 m-2; ideal PEEP: 5.1 [2.1–6.3] l min-1 m-2; P = 0.08). One patient did not complete the protocol because of hypotension.

Conclusion

PEEP of 10–20 cmH2O does not affect gastric mucosal perfusion and is hemodynamically well tolerated in most patients with ARDS, including those receiving adrenergic drugs.

Introduction

Recent studies have shown that lung protective strategies using low tidal volumes and high levels of positive end-expiratory pressure (PEEP) reduce mortality and are becoming standard practice in patients with acute respiratory distress syndrome (ARDS) [1, 2].

Although PEEP improves arterial oxygenation, it can adversely affect systemic hemodynamics, reducing venous return and cardiac output. These effects are proportional to the PEEP level. Regional perfusion can also be affected by PEEP, independently of cardiac output changes. The splanchnic perfusion is particularly sensitive, and any reduction can compromise its barrier function, promote bacterial translocation, and contribute to the development of multiple organ failure [3]. In experimental models, PEEP has markedly decreased mesenteric and portal blood flow, despite only moderate reductions in cardiac output [48]. Similar results have been reported in patients without lung injury [9, 10]. These effects are usually dose related, becoming more pronounced with PEEP levels around 20 cmH2O.

Kiefer reported that PEEP did not significantly alter splanchnic blood flow in six patients with acute lung injury [11]. Nevertheless, caution should be exercised in extending these results to clinical practice, because only hemodynamically stable patients without adrenergic drugs were studied, and PEEP levels never exceeded 14 cmH2O [12].

Our aim was to evaluate the effects of PEEP levels up to 20 cmH2O on gastric mucosal perfusion and systemic hemodynamics in mechanically ventilated patients with ARDS under hemodynamic support.

Methods

Patients

The study was approved by the Ethics Committee of the Medicine Faculty and was performed in the Surgical Intensive Care Unit of the Catholic University Hospital of Chile.

Adult mechanically ventilated patients were considered eligible for the study if they met the following criteria for ARDS during the 24 hours that preceded the study: acute onset of respiratory failure; diffuse bilateral infiltrates in the chest radiograph; a ratio of partial pressure of O2 (PaO2) to fraction of inspired oxygen (FiO2) of less than 200 mmHg; and a pulmonary arterial occlusion pressure less than 18 mmHg and no cardiac failure.

Hemodynamic monitoring included an arterial line and a pulmonary artery catheter (Baxter Edwards Critical-Care, Irvine, CA). Patients could be under vasopressor or inotropic support, but had to be hemodynamically stable for at least 3 hours before starting the protocol.

Patients were excluded if they had any of the following conditions: pregnancy, pre-existing respiratory dysfunction, cardiac index of less than 2.5 l min-1 m-2, or were receiving enteral nutrition.

Interventions

A nasogastric tonometer (TRIP® Tonometry Catheter 14F, with biofilter connector for TONOCAP™ Monitor) was inserted into the stomach and connected to air automated tonometry (TONOCAP™ Monitor; Datex-Engstrom, Helsinki, Finland). All patients were sedated with midazolam and morphine, and paralyzed with rocuronium. Neuromuscular relaxation was measured by a TOF watch® device. An intravenous 20 mg dose of famotidine was administered before starting the protocol. Patients were connected to volume-controlled mechanical ventilation (Servo 900 C; Siemens, Solna, Sweden). A pressure–volume curve was obtained for each patient by the airway occlusion technique [13], and ideal PEEP was defined as the lower inflection point + 2 cmH2O, or 12 cmH2O if no lower inflection point was found.

PEEP levels of 10, 15, 20 cmH2O, and ideal PEEP, with tidal volumes of 8 ml kg-1, were applied in four consecutive 30 min periods, respectively. Respiratory rate was modified to maintain end tidal CO2 within ± 10 mmHg of basal. All patients were receiving a constant infusion of 6% hetastarch before the beginning of the study (40–80 ml h-1). Cardiac output was optimized before and during the trial by determining the respiratory variation of systolic arterial pressure [14]. Whenever the variation was more than 10% a 100 ml bolus of 6% hetastarch was infused and the volume status was reassessed. No change in adrenergic support was allowed during the protocol. If hypotension (mean arterial pressure < 65 mmHg) persisted for more than 1 min, the protocol was stopped.

Measurements

At baseline, and at the end of each period, hemodynamic, ventilatory and tonometric measurements were performed, and arterial blood samples withdrawn. Hemodynamic records included mean arterial pressure, heart rate, cardiac output, pulmonary artery occlusion pressure, central venous pressure and left ventricular stroke work index. Cardiac output was measured by thermodilution as the average of three values obtained after injections of 10 ml of 5% dextrose in water at room temperature. Mean airway pressure, oxygenation index and PEEP levels were registered. Oxygenation index was calculated as mean airway pressure × FiO2 × 100/PaO2. The CO2 gap (gastric partial pressure of CO2 [pCO2] minus arterial pCO2) was calculated by comparing simultaneous measurements of tonometric gastric mucosal pCO2 and arterial pCO2.

Statistical analysis

Results are presented as median and range. The software Statview 5.0 was used to perform the statistical analysis. Nonparametric tests were used because of the small sample size. Data were analyzed with a Friedman test followed by a Wilcoxon signed-rank test if necessary. Results were considered statistically significant at P < 0.05.

Results

Eight patients with ARDS were enrolled. They had a median (range) age of 63.5 years (23–86), and an Acute Physiology and Chronic Health Evaluation II score of 14 (7–20) at admission to the intensive care unit. On the day of the study they had a median Sepsis-related Organ Failure Assessment (SOFA) [15] score of 10 (7–13). All patients fulfilled criteria for ARDS, as defined by the inclusion criteria, during the 24 hours before the study and they had been on mechanical ventilation for 32 (12–72) hours. They were being ventilated with a median PEEP level of 9 (4–12) cmH2O, they had a PaO2/FiO2 ratio of 235 (144–388) mmHg and their respiratory system compliance was 45 (27–60) ml per cmH2O. Seven patients had sepsis (two pneumonia and five extrapulmonary sepsis), and one a severe thoracic trauma. Of the septic patients, six were in septic shock. Characteristics of individual patients are shown in Table 1.

Table 1 Baseline characteristics of the patients

No changes in cardiac index or in CO2 gap were found at any of the study periods (Table 2). Oxygenation index, mean arterial pressure, pulmonary mean arterial pressure, pulmonary artery occlusion pressure, central venous pressure and left ventricular stroke work index also remained stable through the study. Only mean airway pressure and PaO2/FiO2 ratio differed between periods, as expected. Five patients required a 100 ml bolus of hetastarch during the trial; in no patient was it necessary to repeat it. Individual changes in CO2 gap and cardiac index are presented in Figs 1 and 2, respectively. At baseline three patients had already a CO2 gap of more than 20 mmHg. After starting the protocol with 10 cmH2O PEEP, patient 6, who was previously being ventilated with 4 cmH2O PEEP, had a further increase in CO2 gap. When PEEP was increased from 10 to 15 cmH2O, six patients decreased their CO2 gap and two increased it. When PEEP was increased from 15 to 20 cmH2O, three patients increased their CO2 gap, three decreased it and in one patient it remained unchanged. Patient 4 did not complete the protocol because of moderate hypotension (mean arterial pressure 60 mmHg) when PEEP was increased to 20 cmH2O. This patient recovered after an increased dose of norepinephrine (noradrenaline) and a return of PEEP to baseline levels.

Table 2 Respiratory, hemodynamic and tonometric measurements

Six of the eight patients studied survived (75%). The median length of stay in the intensive care unit was 17 (10–34) days and the median duration of mechanical ventilation was 9 (5–34) days.

Figure 1
figure 1

Individual changes in CO2 gap (gastric pCO2 minus arterial pCO2) with different positive end-expiratory pressure levels.

Figure 2
figure 2

Individual changes in cardiac index with different positive end-expiratory pressure levels.

Discussion

Our results show that high PEEP levels (up to 20 cmH2O) do not compromise gastric mucosal perfusion, as assessed by tonometry, and do not affect systemic hemodynamics in most patients with ARDS. This is consistent with the findings of two other studies on the effects of PEEP on splanchnic perfusion in patients with ARDS. Nevertheless, in contrast with our study, neither of those studies included patients in septic shock or under adrenergic support [11, 16].

Shock and cardiovascular dysfunction are frequently associated with ARDS. This is an important issue, because hemodynamic safety concerns could preclude the use of high or optimal PEEP levels in that setting, even if necessary. A major finding of our study is that PEEP levels up to 20 cmH2O can be well tolerated, even in patients with ARDS and septic shock. Nevertheless, our trial was relatively short and we cannot exclude the possibility that keeping high PEEP levels for a longer period might result in increased fluid requirements, which could be deleterious in the longer term.

Experimental and clinical research has demonstrated that in mechanically ventilated subjects without lung injury, PEEP decreases venous return and, secondarily, cardiac output [1719]. In addition, Trager and colleagues showed that, in patients with acute respiratory failure associated with septic shock, high PEEP levels induced a decrease in cardiac output [20]. In contrast, we found no decrease in cardiac output in our patients tested with increasing PEEP levels when fluid administration was optimized according to respiratory variation in systolic arterial pressure. A similar result was reported by Kiefer and colleagues and by Akinci and colleagues [11, 16]. Possible explanations for these contradictory results are a higher rate of fluid administration and the use of lower tidal volumes in the latter studies. Although we did not determine the upper inflection point of the pressure–volume curve, we think that by keeping tidal volume at 8 ml kg-1 any overdistension of the lungs was minimized. Lung volumes are a critical component of the hemodynamic effects of ventilation [21]. Thus, it seems that it is possible to preserve cardiac output in patients with ARDS, despite the use of high PEEP levels, by optimizing fluid administration and limiting tidal volumes.

Gastric mucosal perfusion, as assessed by CO2 gap, also remained unchanged during the PEEP trial. This is consistent with the results reported by Kiefer and Akinci in similar studies. In all these studies cardiac output remained unchanged [11, 16]. In contrast, Trager reported, in a series of septic shock patients with acute respiratory failure, that an increase in PEEP from 5 to 15 cmH2O induced a decrease in cardiac output associated to a decrease in hepatic vein O2 saturation and in hepatic glucose production [20]. It therefore seems that by avoiding decreases in cardiac output, splanchnic perfusion can be preserved in the majority of the patients.

In spite of the fact that no significant changes in CO2 gap or cardiac index were found during the protocol, when looking at the individual data certain patients evidenced an adverse effect when their PEEP level was increased. Patient 4, who had an extrapulmonary ARDS, presented hypotension when 20 cmH2O PEEP was applied. In this case, no simultaneous records of cardiac output or CO2 gap could be made for safety reasons (we immediately proceeded to decrease PEEP level). Patient 6, who had a pulmonary ARDS and who before starting the study had a 30 mmHg CO2 gap while being ventilated with 4 cmH2O PEEP, presented a further deterioration in CO2 gap after starting the protocol with 10 cmH2O PEEP, which was not associated with a decrease in cardiac output. Thereafter, the CO2 gap remained unchanged despite increasing PEEP up to 20 cmH2O. These events suggest that not all patients with ARDS can tolerate high PEEP levels; if required, careful hemodynamic monitoring including assessment of regional perfusion should be applied.

One major limitation of our study is the small number of patients studied. Thus, a type II error cannot be excluded. We did not perform any a priori power analysis because we had no estimation of the possible magnitude of the effects that PEEP could have on gastric tonometry.

Another limitation is the rather moderate severity of ARDS in our study. Although all patients fulfilled criteria for ARDS during the 24 hours that preceded the study, at inclusion their PaO2/FiO2 ratio and their respiratory system compliance were only moderately decreased. Two recent papers provide an explanation for this observation [22, 23]. They show in patients diagnosed with ARDS that after a few hours of treatment with PEEP or a high FiO2, more than half of the patients present a PaO2/FiO2 ratio of more than 200 mmHg. In addition, respiratory system compliance increased by more than 10 ml per cmH2O after 6 hours of treatment with PEEP [23]. At inclusion our patients had already been ventilated with a median PEEP level of 9 cmH2O for more than 12 hours, which could have explained the rather improved respiratory performance at baseline. In any event, this improvement demonstrated a less severe ARDS. It is possible that more severely compromised patients might present a lower tolerance to high PEEP levels.

Other limitation is that tonometry was the sole method used to assess gastric mucosal perfusion. Nevertheless, Elizalde and colleagues showed that gastric mucosal blood flow, measured by laser Doppler flowmetry and by reflectance spectrophotometry, is well correlated with gastric intramucosal acidosis in mechanically ventilated patients [24].

Conclusions

Our study supports the findings of previous studies suggesting that high PEEP levels do not affect splanchnic perfusion and are hemodynamically well tolerated in most patients with ARDS. Furthermore, our study shows that gastric mucosal perfusion can be well preserved while high PEEP levels are applied even in patients presenting cardiovascular dysfunction and receiving adrenergic support, which is a frequent occurrence in critical care. Nevertheless, two of the eight patients studied exhibited adverse effects during the PEEP trial, which highlights the importance of a close monitoring of systemic and regional perfusion while applying high PEEP levels to patients with ARDS. Future studies should assess the effects of PEEP on splanchnic perfusion in a longer term.

Key messages

  • High PEEP levels do not affect gastric mucosal perfusion and are hemodinamically well tolerated in most patients with ARDS

Abbreviations

ARDS:

acute respiratory distress syndrome

CO2 gap:

gastric pCO2 minus arterial pCO2

FiO2 :

fraction of inspired oxygen

FiO2 :

partial pressure of O2

pCO2 :

partial pressure of CO2

PEEP:

positive end-expiratory pressure.

References

  1. The Acute Respiratory Distress Syndrome Network: Ventilation with lower tidal volumes as compared with traditional tidal volumes for acute lung injury and the acute respiratory distress syndrome. N Engl J Med 2000, 342: 1301-1308. 10.1056/NEJM200005043421801

    Article  Google Scholar 

  2. Amato MB, Barbas CS, Medeiros DM, Magaldi RB, Schettino GP, Lorenzi-Filho G, Kairalla RA, Deheinzelin D, Munoz C, Oliveira R, Takagaki TY, Carvalho CR: Effect of a protective-ventilation strategy on mortality in the acute respiratory distress syndrome. N Engl J Med 1998, 338: 347-354. 10.1056/NEJM199802053380602

    Article  CAS  PubMed  Google Scholar 

  3. Gutierrez G, Palizas F, Doglio G, Wainsztein N, Gallesio A, Pacin J, Dubin A, Schiavi E, Jorge M, Pusajo J, Klein F, San Roman E, Dorfman B, Shottlender J, Giniger R: Gastric intramucosal pH as a therapeutic index of tissue oxygenation in critically ill patients. Lancet 1992, 339: 195-199. 10.1016/0140-6736(92)90002-K

    Article  CAS  PubMed  Google Scholar 

  4. Love R, Choe E, Lippton H, Flint L, Steinberg S: Positive end-expiratory pressure decreases mesenteric blood flow despite normalization of cardiac output. J Trauma 1995, 39: 195-199.

    Article  CAS  PubMed  Google Scholar 

  5. Fournell A, Scheeren TW, Schwarte LA: PEEP decreases oxygenation of the intestinal mucosa despite normalization of cardiac output. Adv Exp Med Biol 1998, 454: 435-440.

    Article  CAS  PubMed  Google Scholar 

  6. Fujita Y: Effects of PEEP on splanchnic hemodynamics and blood volume. Acta Anaesthesiol Scand 1993, 37: 427-431.

    Article  CAS  PubMed  Google Scholar 

  7. Lehtipalo SBB, Arnelov C, Frojse R, Johansson G, Winso O: PEEP can induce splanchnic ischemia during critical reductions in regional perfusion pressure. Intensive Care Med 2000, 26: S375.

    Google Scholar 

  8. Arvidsson D, Almquist P, Haglund U: Effects of positive end-expiratory pressure on splanchnic circulation and function in experimental peritonitis. Arch Surg 1991, 126: 631-636.

    Article  CAS  PubMed  Google Scholar 

  9. Berendes E, Lippert G, Loick HM, Brussel T: Effects of positive end-expiratory pressure ventilation on splanchnic oxygenation in humans. J Cardiothorac Vasc Anesth 1996, 10: 598-602.

    Article  CAS  PubMed  Google Scholar 

  10. Aneman A, Eisenhofer G, Fandriks L, Olbe L, Dalenback J, Nitescu P, Friberg P: Splanchnic circulation and regional sympathetic outflow during peroperative PEEP ventilation in humans. Br J Anaesth 1999, 82: 838-842.

    Article  CAS  PubMed  Google Scholar 

  11. Kiefer P, Nunes S, Kosonen P, Takala J: Effect of positive end-expiratory pressure on splanchnic perfusion in acute lung injury. Intensive Care Med 2000, 26: 376-383. 10.1007/s001340051170

    Article  CAS  PubMed  Google Scholar 

  12. De Backer D: The effects of positive end-expiratory pressure on th splanchnic circulation. Intensive Care Med 2000, 26: 361-363. 10.1007/s001340051168

    Article  CAS  PubMed  Google Scholar 

  13. Levy PST, Corbeil C, Albala M, Pariente R, Milic-Emili J, Jonson B: A method for studying the static volume-pressure curves of the respiratory system during mechanical ventilation. J Crit Care 1989, 4: 83-89. 10.1016/0883-9441(89)90122-6

    Article  Google Scholar 

  14. Gunn SR, Pinsky MR: Implications of arterial pressure variation in patients in the intensive care unit. Curr Opin Crit Care 2001, 7: 212-217. 10.1097/00075198-200106000-00012

    Article  CAS  PubMed  Google Scholar 

  15. Vincent JL, Moreno R, Takala J, Willatts S, De Mendonca A, Bruining H, Reinhart CK, Suter PM, Thijs LG: The SOFA (Sepsis-related Organ Failure Assessment) score to describe organ dysfunction/failure. On behalf of the Working Group on Sepsis-Related Problems of the European Society of Intensive Care Medicine. Intensive Care Med 1996, 22: 707-710. 10.1007/s001340050156

    Article  CAS  PubMed  Google Scholar 

  16. Akinci IO, Cakar N, Mutlu GM, Tugrul S, Ozcan PE, Gitmez M, Esen F, Telci L: Gastric intramucosal pH is stable during titration of positive end-expiratory pressure to improve oxygenation in acute respiratory distress syndrome. Crit Care 2003, 7: R17-R23. 10.1186/cc2172

    Article  PubMed Central  PubMed  Google Scholar 

  17. Dorinsky PM, Whitcomb ME: The effect of PEEP on cardiac output. Chest 1983, 84: 210-216.

    CAS  PubMed  Google Scholar 

  18. Liebman PR, Patten MT, Manny J, Shepro D, Hechtman HB: The mechanism of depressed cardiac output on positive end-expiratory pressure (PEEP). Surgery 1978, 83: 594-598.

    CAS  PubMed  Google Scholar 

  19. Pick RA, Handler JB, Murata GH, Friedman AS: The cardiovascular effect of positive end-expiratory pressure. Chest 1982, 82: 345-350.

    Article  CAS  PubMed  Google Scholar 

  20. Trager K, Radermacher P, Georgieff M: PEEP and hepatic metabolic performance in septic shock. Intensive Care Med 1996, 22: 1274-1275.

    Article  CAS  PubMed  Google Scholar 

  21. Pinsky MR: Recent advances in the clinical application of heart–lung interactions. Curr Opin Crit Care 2002, 8: 26-31. 10.1097/00075198-200202000-00005

    Article  PubMed  Google Scholar 

  22. Ferguson ND, Kacmarek RM, Chiche JD, Singh JM, Hallett DC, Mehta S, Stewart TE: Screening of ARDS patients using standardized ventilator settings: influence on enrollment in a clinical trial. Intensive Care Med 2004, 30: 1111-1116. 10.1007/s00134-004-2163-2

    Article  PubMed  Google Scholar 

  23. Estenssoro E, Dubin A, Laffaire E, Canales HS, Saenz G, Moseinco M, Bachetti P: Impact of positive end-expiratory pressure on the definition of acute respiratory distress syndrome. Intensive Care Med 2003, 29: 1936-1942. 10.1007/s00134-003-1943-4

    Article  PubMed  Google Scholar 

  24. Elizalde JI, Hernandez C, Llach J, Monton C, Bordas JM, Pique JM, Torres A: Gastric intramucosal acidosis in mechanically ventilated patients: role of mucosal blood flow. Crit Care Med 1998, 26: 827-832. 10.1097/00003246-199805000-00011

    Article  CAS  PubMed  Google Scholar 

Download references

Author information

Authors and Affiliations

Authors

Corresponding author

Correspondence to Alejandro Bruhn.

Additional information

Competing interests

None declared.

Rights and permissions

Reprints and permissions

About this article

Cite this article

Bruhn, A., Hernandez, G., Bugedo, G. et al. Effects of positive end-expiratory pressure on gastric mucosal perfusion in acute respiratory distress syndrome. Crit Care 8, R306 (2004). https://doi.org/10.1186/cc2905

Download citation

  • Revised:

  • Accepted:

  • Published:

  • DOI: https://doi.org/10.1186/cc2905

Keywords