Tidal volume accuracy during non-invasive ventilation with modern neonatal mechanical ventilators

Introduction

Maintaining the appropriate tidal volume (VT) is important for success of lung protective ventilation. However, in neonates, the presence of airway leaks may increase the errors in the delivery of tiny VT, which raises a concern of ventilator-induced lung injury. This study is to investigate the accuracy of VT delivery during non-invasive ventilation (NIV) with modern neonatal ventilators.

Methods

Using a lung simulator for a patient body weight of 3 kg, we measured the actual delivered VT in the lung and compared it with the value displayed on the ventilator in six ventilators. We tested 18 conditions with various combinations of respiratory mechanics (normal, restrictive, obstructive), leak levels (0, 1.0, 1.5 l/minute), and PEEP settings (5, 10 cmH2O). All conditions were tested in NIV mode. The pressure level was set to achieve VT to the lung at 6 to 7 ml/kg. All other settings were: F1O2 0.21, Itime 0.6 seconds, f 25/minute, and default rise time. We calculated the mean errors of the ventilator-displayed VT at various levels of airway leak.

Results

The VT mean error values are presented in Table 1. When no leak existed, the mean error was less than 5% in all ventilators except one (C3) which showed a mean error of 26%. As the leak level increased, three ventilators (C3, G5, and VN500) showed marked differences between the delivered and displayed VT. In particular, the VN500 could not operate in the large leak condition. The other three ventilators (PB840, PB980, Servo i) showed acceptable VT accuracy across all conditions tested.

Table 1

Conclusion

Tidal volume accuracy during neonatal NIV varies greatly among different ventilators and leak conditions. This must be considered in neonatal ventilation management to avoid overventilation or underventilation.

Author information

Authors and Affiliations

Authors

Rights and permissions

Reprints and permissions

About this article

Check for updates. Verify currency and authenticity via CrossMark

Cite this article

Moon, K., Mizuguchi, S., Tachibana, K. et al. Tidal volume accuracy during non-invasive ventilation with modern neonatal mechanical ventilators. Crit Care 19 (Suppl 1), P262 (2015). https://doi.org/10.1186/cc14342

Download citation

  • Published:

  • DOI: https://doi.org/10.1186/cc14342

Keywords