From: Tidal volume accuracy during non-invasive ventilation with modern neonatal mechanical ventilators

Leak amount (l/minute)

Hamilton C3

Hamilton G5

Servo i

PB840

PB980

VN500

0

26 (15)

4 (6)

3 (2)

2 (7)

-4 (2)

2 (2)

1.0

-9 (5)

9 (5)

-2 (2)

-5 (4)

-2 (4)

20 (5)

1.5

16 (8)

-14 (8)

-2 (2)

-6 (4)

-2 (3)

33 (6)a

  1. Data presented as mean (SD), %. aUnable to measure in one case.