Methods

In this study, five NIV trials were randomly applied, preceded and followed by a trial of unassisted spontaneous breathing (SB). Baseline settings, for example, 5 cmH2O of both inspiratory pressure support (PS) and positive end-expiratory pressure (PEEP), were applied through FM, HS and HN, while increased settings (PS and PEEP of 8 cmH2O) were only applied through HS and HN. We measured flow, airway, esophageal and gastric pressures, and calculated inspiratory effort indexes and trigger delays. Comfort was assessed with a visual-analog-scale.

Results

We found that FM, HS and HN at baseline settings were not significantly different with respect to inspiratory effort indexes and comfort. Inspiratory trigger delay and time of synchrony (TI,synchrony) were significantly improved by FM compared to both helmets, whereas expiratory trigger delay was shorter with FM, as opposed to HS only. HN at increased settings performed better than FM in decreasing inspiratory effort measured by pressure-time product of transdiaphragmatic pressure (PTPdi)/breath (10.7 ± 9.9 versus 17.0 ± 11.0 cmH2O*s), and PTPdi/min (128 ± 96 versus 204 ± 81 cmH2O*s/min), and PTPdi/L (12.6 ± 9.9 versus 30.2 ± 16.8 cmH2O*s/L). TI, synchrony was inferior between HN and HS at increased settings and FM.

Conclusions

HN might hold some advantages with respect to interaction and synchrony between subject and ventilator, but studies on patients are needed to confirm these findings.

Trial registration

ClinicalTrials.gov NCT01610960

Introduction

The choice of the interface is one of the crucial factors determining the success of noninvasive ventilation (NIV) in both the acute [1] and chronic [2] setting. Different types of face mask (FM) that is, oral, nasal, oronasal, total full face of different size, design, and material are available to increase the patient’s comfort. Despite the broad availability of FMs, most NIV failures are still associated with mask-related side effects such as air leaks [3], skin breakdown [4, 5], and mask discomfort [1, 2, 6].

New interfaces have been introduced with the aim of improving patient comfort and overcoming these side effects. The helmet is a transparent latex-free polyvinylchloride hood; in its standard configuration it is joined by a rigid plastic ring to a soft collar and secured by two padded armpit braces at four hooks (two in the front and two in the back of the plastic ring). The standard helmet (HS) has been shown to improve NIV comfort over time compared to FM and to reduce skin breakdown, gastric distension, and eye irritation [7, 8]. Despite these advantages, the helmet has specific drawbacks primarily related to its highly compliant soft collar [9]. Inspiratory pressure dissipation increases the time lag between inspiratory effort and ventilator assistance, worsening patient ventilator synchrony [1012]. Furthermore, compared to FM the helmet has been found to be associated with less efficient inspiratory-muscle unloading, both in healthy volunteers [11] and patients [10, 12]. In addition, the armpit braces that maintain the helmet in place may cause patients discomfort and skin lesions leading to NIV intolerance and failure [13].

To reduce these technical problems, a new helmet (HN) (NIV-Castar R Next, Starmed, Mirandola, Italy) has been developed and introduced into clinical use, in which an opening ring placed underneath an inflatable cushion secures the helmet without the need for armpit braces. Moreover, compared to HS, with HN the pressure dissipated because of the downward displacement of the soft collar during ventilator insufflation is eliminated, or at least reduced to a large extent. Very recently a bench study comparing HS and HN suggested a better performance of the latter in terms of triggering performance, patient-ventilator synchrony, and rate of pressurization [14].

The rotational use of different interfaces has been proposed to improve patients’ tolerance and prolong NIV application both in hypoxemic [15] and hypercapnic [16] patients. On one hand, patients needing several daily hours of NIV may benefit from the rotational use of different interfaces to reduce the risk of discomfort and the side effects of one specific interface [1517]; on the other hand, the use of diverse interfaces with equal ventilator assistance may result in a different extent of muscle unloading and patient-ventilator synchrony [10, 12]. To overcome the decreased efficacy of the HS in delivering ventilator assistance, compared to the FM, Vargas et al. recently proposed to increase the level of both positive end-expiratory pressure (PEEP) and inspiratory pressure support (PS) [12].

The objective of this physiologic study was to assess and compare the efficacy of FM, HS, and HN in delivering NIV in healthy volunteers. Interaction and synchrony between subject and ventilator, and the subject’s inspiratory effort and comfort were compared at baseline settings, that is, at PEEP 5 cmH2O and PS 5 cmH2O, and at increased settings, that is, PEEP 8 cmH2O and PS 8 cmH2O.

Materials and methods

The protocol was approved by the institutional review board (Comité de Protection des Personnes Sud Méditéranée IV, Montpellier; approval number 2012-A00098-35), and written informed consent was obtained from each subject. Twelve healthy volunteers were studied in a semi-recumbent position. The subjects (three women and nine men, age 30 ± 9 years) were all nonsmokers and had a body mass index of 24 ± 4 kg/m2. NIV was delivered trough an FM (Performatrak, Philips Respironics), HS (NIV-Castar R), and HN (NIV-Castar R Next), using an ICU ventilator equipped with software for air-leaks compensation (NIV module) (Servo-I, Maquet, Solna, Sweden) set in PS ventilation mode.

Measurements

Airflow was measured with a pneumotachograph (Fleish no. 2; Fleisch, Lausanne, Switzerland) connected to a differential pressure transducer (MP45, ± 2 cmH2O; Validyne, Northridge, CA, USA). The pneumotachograph was placed between the mask inlet and the Y-piece of the ventilator circuit during FM, and distally to the mouthpiece during both spontaneous breathing and helmet ventilation as shown in Figure 1. Tidal volume (VT) was obtained by digital integration of the flow signal. Pressure at the airway opening (Pao) was assessed via a side port connected to a pressure transducer. Esophageal (Pes) and gastric (Pga) pressures were measured using a double balloon-tipped catheter (Nutrivent, Sidam s.r.l., Mirandola, Italy) positioned in the mid-esophagus and in the stomach and connected to two differential pressure transducers. The esophageal and gastric balloons were inflated with 4 ml of air. The correct positioning of the esophageal balloon was checked by an inspiratory occlusion [18]. Adequate placement of the gastric balloon was ascertained by gentle manual pressure on the subject’s abdomen to observe fluctuations in gastric pressure, as well as by asking the patient to swallow and verifying that the sharp increase in the esophageal pressure caused by esophageal contraction was not observed on the Pga tracing. Transdiaphragmatic (Pdi) pressure was obtained by subtracting Pes from Pga.

Figure 1
figure 1

Schematic representation of the experimental design. (A) Spontaneous breathing; (B) face mask; (C) helmet standard; (D) helmet next. Pao, pressure at the airway opening; Pes, esophageal pressure; Pga, gastric pressure.

Patient’s own (neural) respiratory rate (RRn) and ventilator rate of cycling (RRv) were determined from Pdi and Pao swings, respectively. In line with previous studies [10, 11, 19, 20], the patient’s inspiratory time (TI,p) was estimated from the Pdi tracing as the time between the onset of the positive Pdi swing above baseline (that is, the start of inspiratory effort) and the point where Pdi started to fall toward baseline. Likewise, the duration of the inspiratory assistance provided by the ventilator (TI,v) was calculated from the Pao tracing. The inspiratory trigger delay (Delay,TR-insp) was calculated as the time lag between onset of inspiratory effort and start of ventilator support, and the expiratory trigger delay (Delay,TR-exp) was calculated as the time lag between the points at which Pdi and Pao started to fall toward baseline. We also calculated the time of synchrony between muscle effort and ventilator support (TI,p-TI,v synchrony) as the period in the course of inspiration during which the diaphragm was contracting and the ventilator was concurrently delivering support.

The pressure-time product of the transdiaphragmatic (PTPdi) pressure was assessed to determine the effort exerted by the diaphragm and to provide a surrogate estimate of the overall inspiratory muscle exertion, not considering the amount of effort spent to distend the chest wall, from the changes in Pdi over time. The pressure-time product was calculated per minute, determining the area under Pdi (PTPdi/min) within a 1-minute time interval, per breath (PTPdi/br), dividing PTPdi/min by RRn, and per liter, dividing PTPdi/min by minute ventilation (VE) [10, 12, 21].

The inspiratory work of breathing (WOB) performed by the patient was computed from Pes and VT loops as previously described [22, 23]. Briefly, the inspiratory work per breath was calculated from the Campbell diagram by computing the area enclosed between the inspiratory esophageal pressure-VT curve on the one hand, and the static esophageal pressure-volume curve of the chest wall on the other hand, using a theoretical value for chest-wall compliance (4% of the predicted value of the vital capacity per cm H2O). Although we directed careful attention to minimizing leaks around the mask, this problem did occur and was taken into consideration. Because leaks around the mask are more likely to occur during inspiration, leading to overestimation of the inspired volume, we calculated inspiratory WOB by applying a correction factor to the inspiratory flow based on the patient’s expired minute volume [23, 24]. For this, we measured the ratio of expired over inspired minute volume and applied a correction factor equal to this ratio to the flow signal used to measure inspiratory WOB. Inspiratory WOB was expressed as the work per breath (J/breath), as the work per volume unit (J/L), and as the work per time unit (J/min).

Comfort was assessed at the end of each trial, using a 0- to 10-item visual analog scale [25, 26] (0 = intolerable discomfort, 3 = strong discomfort, 8 = moderate discomfort, 10 = no discomfort). Heart rate and pulse arterial oxygen saturation were monitored throughout the protocol.

Experimental protocol

A trial of spontaneous unassisted breathing (SB) was performed in all subjects at the beginning of the study protocol (SBI), followed by three NIV trials with FM, HS, HN at baseline settings (PEEP and PS set at 5 cmH2O) and two additional trials with the increased settings (PS and PEEP set at 8 cmH2O) for HS and HN only. In all five NIV conditions, each period lasted 5 minutes. The interfaces were randomly applied following a computer-generated random sequence. Once the interface was determined a second computer-generated random sequence was applied to establish the setting. To conclude, a second trial of spontaneous unassisted breathing was performed (SBE). A schematic representation of the experimental setup used during unassisted spontaneous breathing, FM, HS and HN ventilation is shown in Figure 1, and the new helmet is depicted in detail in Figure 2.

Figure 2
figure 2

Helmet next. The new helmet (Helmet next) is composed of a rigid plastic ring (1), inflatable cushion (2) with its inflating line (3), anular opening ring secured to the rigid ring underneath the cushion (4), and transparent hood (5). The helmet is secured to the head by the inflated cushion placed around the neck below the jaw.

The ICU ventilator Servo-i was set in NIV mode with an inspired oxygen fraction (FiO2) of 21%, the fastest rise time, an inspiratory trigger at five units, corresponding to one l/minute, and cycling-off flow threshold of 25% of the peak inspiratory flow. These last settings were maintained unmodified throughout the study.

During the SB, the HS, and HN trials, the subjects wore a nose clip and breathed through a mouthpiece connected to a side port connector for the measurement of the pressure at the airway opening, and to a pneumotachograph. The measuring equipment had a dead space of 49 ml and resistance of 2.0 and 2.6 cmH2O L-1 s-1 at a flow rate of 0.5 and 1.0 L/s, respectively.

Data analysis

After elimination of cycles contaminated by artifacts due to cough and esophageal spasms, 20 consecutive breaths, closest to the end of the 5-minute run, were used to compute average values. All observations were included in the analysis. The signals were amplified, low-pass filtered, digitized at 128 Hz and sampled using an analog-to-digital converter system (MP100; Biopac Systems, Santa Barbara, CA, USA). Results are expressed as means ± SD and as group-mean difference ± SD. Normality of quantitative variables (Shapiro-Wilk test) was checked. Data were compared using a generalized linear mixed-effects model for repeated measures, taking into account repeated measures as random effects, and NIV condition, cross-over sequence and its interaction as fixed effects. Normality of residuals was checked after applying the linear mixed-effects model. Comparisons defined a priori were FM versus HS, FM versus HN, SBI versus HS, and SBI versus HN. Significance was set at P <0.0125 after correction using the Bonferroni t-test for the number of multiple comparisons, that is, four.

Results

Vital signs and comfort

No subjects asked to discontinue the trial for any reason. Heart rate and pulse arterial oxygen saturation remained unchanged, regardless of interface and ventilator setting (data not shown). Increasing Pao produced a progressive decrease in patient comfort that wasd statistically significant only between SB, FM and increased settings. In fact, comfort was scored 8.9 ± 1.3 during SB and 7.7 ± 1.1, 7.5 ± 1.4, and 7.4 ± 1.8 at baseline settings during FM, HS, and HN, respectively, whereas the scores were 6.4 ± 1.6 and 6.7 ± 1.6 at increased settings for HS and HN, respectively. Group-mean differences between FM and HS, FM and HN, SB and HS, SB and HN for comfort were 1.4 ± 1.1, 1.0 ± 1.4, 2.5 ± 1.3, and 2.2 ± 1.2, respectively. Of note, there was no significant difference between the interfaces on iso-assistance.

Table 1 presents the baseline data and Table 2 the data following intervention at increased settings, with group-mean differences between FM and SB versus HS and HN presented as mean ± SD (see Table E1 in Additional file 1 in which data are presented as median (interquartiles) in the appendix).

Table 1 Main ventilatory variables and inspiratory efforts at baseline
Table 2 Main ventilatory variables and inspiratory effort following interventions, at increased settings, with group means differences between FM and SB versus HS and HN

Breathing pattern and inspiratory effort

An example of illustrative traces, that is, Pao, flow, Pes, Pga, and Pdi recorded during the experimental trials, is depicted in Figures 3 and 4.

Figure 3
figure 3

Representative traces of a patient (patient 1) during the randomized trials for spontaneous breathing, face mask at PEEP and PS of 5 cmH 2 O, helmet standard at PEEP and PS of 5 cmH 2 O, helmet standard at PEEP and PS of 8 cmH 2 O, helmet next at PEEP and PS of 5 cmH 2 O, helmet next at PEEP and PS of 8 cmH 2 O, and spontaneous breathing. Pao, Flow, Pes, Pga, and Pdi are shown from top to bottom. FM, face mask; HN, helmet next; HS, helmet standard; Pao, pressure at the airway opening; Pdi, transdiaphragmatic pressure; PEEP, positive end-expiratory pressure; Pe, esophageal pressure; Pga, gastric pressure; PS, pressure support.

Figure 4
figure 4

The pressure-time product of the transdiaphragmatic pressure calculated per breath, per minute, and per liter. (A) Calculation per breath; (B) calculation per minute; (C) calculation per liter. FM: face mask; HN: helmet next; HS: helmet standard; PEEP: positive end-expiratory pressure; PS: pressure support; PTPdi: pressure-time product of the transdiaphragmatic pressure; SB: spontaneous breathing.

Expiratory VT was lower, with FM, as opposed to all other conditions. The differences in the respiratory rate (RR) were small and not significant. Minute ventilation (VE) significantly increased with HN at baseline settings (P <0.0125) and with both helmets at increased settings (P <0.0125) compared to FM.

Compared with SBI, NIV administered at baseline settings did not modify inspiratory effort, as assessed by pressure-time product of PTPdi/breath, PTPdi/min and PTPdi/L, irrespective of the interface applied. HN at increased settings performed better than FM in decreasing inspiratory effort measured by PTPdi/breath (10.7 ± 9.9 versus 17.0 ± 11.0 cmH2O*s), PTPdi/min (128 ± 96 versus 204 ± 81 cmH2O*s/min), and PTPdi/L (12.6 ± 9.9 versus 30.2 ± 16.8 cmH2O*s/L). There was no significant difference in WOB irrespective of support level and type of interface applied.

Patient-ventilator synchrony

At baseline settings, Delay,TR-insp was significantly shorter (P <0.0125) with FM (0.181 ± 0.086 s), compared to both HS and HN (0.354 ± 0.081 s and 0.276 ± 0.091 s, respectively). At increased settings, Delay,TR-insp was shorter (P <0.0125) with HN (0.230 ± 0.061 s), compared to HS (0.345 ± 0.073 s). With both HS and HN the increased settings decreased Delay,TR-insp, which, however, was not different to FM with HN only.

At baseline settings, Delay,TR-exp was similar between FM (0.275 ± 0.038 s) and HN (0.301 ± 0.044 s), whereas for HS (0.395 ± 0.093 s) it was significantly longer (P <0.0125). Although at increased settings Delay,TR-exp slightly improved with both HN (0.290 ± 0.099 s) and HS (0.367 ± 0.074 s), with the latter it remained significantly longer, as opposed to FM.

Time of synchrony (TI,synchrony), expressed as percent of patient’s inspiratory time, was longer during FM (90 ± 4%), compared to HS (79 ± 7%) (P <0.0125) and HN (83 ± 4%) at baseline settings. At increased settings, no difference was found between helmets, but compared to FM, TI,synchrony remained significantly lower with HS (80 ± 5%) and HN (85 ± 2%).

Discussion

We found that in healthy volunteers FM, HS and HN at baseline settings, that is, PEEP 5 cmH2O and PS 5 cmH2O, were not significantly different with respect to inspiratory muscles unloading, WOB and comfort. With respect to patient-ventilator interaction, TI,synchrony and Delay,TR-insp were significantly improved by FM compared to both helmets, whereas Delay,TR-exp was shorter with FM, as opposed to HS, but not HN. At increased settings (that is, PEEP 8 cmH2O and PS 8 cmH2O), inspiratory effort was significantly lower, compared to SBI and FM, with HN, while not HS. Comfort was inversely related to the pressure applied to the airway opening by the ventilator, regardless of the interface. Patient-ventilator synchrony, as assessed by Delay,TR-insp and Delay,TR-exp, was equivalent between FM at baseline setting and HN at increased settings, but remained inferior with HS even at increased settings. TI synchrony remained inferior with both helmets at increased settings, compared to FM.

Before discussing our findings, some limitations of this study merit consideration. First, our investigation was conducted on healthy individuals, which makes our findings not entirely inferable to the clinical setting. We share this limit with several studies, some of which were performed to evaluate HS at the time it was introduced into clinical use [9, 11, 21]. As a matter of fact, studies on HN are lacking, which justifies this first in vivo evaluation in healthy volunteers. Second, of the seven trials performed on each subject, the two spontaneous breathing trials and four helmet trials had the same breathing apparatus with mouth-piece and nose-clip, whereas the FM trial did not. The use of breathing apparatus that is, mouth-piece and nose-clip, has been repeatedly shown to affect breathing pattern by increasing VT and VE with few changes in the RR [27, 28]. In keeping with these studies, during the FM trial VT and VE were smaller than during the other six trials, the reduction in VE being statistically significant, whereas the RR was only slightly decreased. As VT is one of the variables, in addition to Pes, computed in the measurement of WOB, the use of a breathing apparatus might have influenced it to a similar extent. Third, we limited the time of each trial to 5 minutes, which is indeed rather short, especially when compared to that utilized in other similar studies performed either in healthy volunteers [9, 21] or patients [10, 12]. It should be considered, however, that in contrast to these studies, in our experimental set up the breathing apparatus was connected to the subject through the mouthpiece, which produces discomfort, potentially affecting either the breathing pattern or the indexes of respiratory muscle effort. Although we cannot exclude that the relatively short period of evaluation could be not sufficient to fully achieve a steady state, we are convinced that the poor tolerance to the breathing apparatus would have represented more bias. Finally, potential bias related to the non-blinded setting of the study has to be taken into consideration, although we share this limitation with several studies [9, 11, 21].

In hypercapnic patients with severe chronic obstructive pulmonary disease, Navalesi et al. observed a significant decrease per minute in PTPdi of 65% and 43% with FM and HS, respectively, as opposed to SB [10, 12]. In patients at high risk of developing post-extubation respiratory failure, Vargas et al. found that at baseline settings, FM and HS significantly reduced PTPdi per minute compared to SB, at 69% and 60%, respectively [12]. In our healthy volunteers, we also observed a reduction in PTPdi per minute at baseline settings with respect to SB, but the extent of this decrease range was lower (15 to 20%) and not statistically significant, or significantly different between the three interfaces. On one hand, these discrepancies are likely a consequence of the lower inspiratory pressure applied and on the other, due to the fact that in contrast to the aforementioned studies [10, 12], we performed our investigation in normal subjects without underlying respiratory disorders. Indeed, other previous studies on healthy subjects at similar baseline settings also failed to demonstrate a difference in PTPdi between FM and HS[11], and between HS and SB [21]. In keeping with Vargas et al., we found that at increased settings HS and HN were both as effective as FM at the baseline setting, in reducing inspiratory effort compared to SB; of note, PTPdi expressed per breath, per minute or per liter, was significantly lower with HN at increased settings than with FM at baseline settings.

Surprisingly, the differences in WOB between trials were rather small and not significant. The large VT, that to some extent was a consequence of the breathing apparatus [27, 28] as already mentioned, may have influenced this variable, which is influenced by both Pes and VT. This may also explain the lower, though not significantly lower values of WOB/breath and WOB/minute observed during the FM trial, in contrast to WOB/L, which, in fact was not influenced by the amount of VT. A possible expiratory activation of the expiratory respiratory muscles may contribute to the better performance of both HS and HN compared to the FM.

In agreement with previous work [10, 12], at baseline settings both Delay,TR-insp and Delay,TR-exp were longer with HS than with FM, whereas with HN Delay,TR-insp, but not Delay,TR-exp, was significantly higher, with respect to FM. Whereas Vargas et al. found that Delay,TR-insp with HS was significantly lower with the increased settings than with the baseline settings, in our study HS at increased settings compared to HS at baseline settings did not produce significant improvements in Delay,TR-insp, Delay,TR-exp or TI,synchrony, which all remained worse than the corresponding values obtained with FM at standard settings. Conversely, with HN the increased settings improved all three variables, which were not significantly different compared to FM at baseline.

HS was shown to be better tolerated than FM in studies comparing the two interfaces over the medium (hours) and long (days) period [7, 8]. In the present investigation, comfort was not influenced by the type of interface, as already reported in the previous short term physiologic investigations comparing FM and HS over brief periods of time (20 to 30 minutes) [10, 12], which has been attributed to the longer time required for most of the major determinants of patient discomfort during NIV to take place [10, 12]. Consistent with previous work on healthy volunteers [21], however, we found that comfort was affected by the amount of applied pressure.

Very recently, studies conducted either in hypoxemic [15] and hypercapnic [16] patients, and a European survey [17] on the use of NIV, indicate the rotational use of different interfaces as a possible means to improve patients’ tolerance and prolong NIV application. Our study shows that different interfaces can be used to achieve comparable results with respect to inspiratory effort reduction. In addition, we found that in healthy volunteers, although FM performs better overall than helmets at baseline settings, at increased settings both helmets had improved performance and resulted in lower inspiratory effort lower than SB with both helmets, and even better than FM with HN only. Furthermore, at increased settings HN achieved better results for interaction and synchrony between subject and ventilator compared to HS, which indicated that this new interface performed similarly to FM.

Conclusions

HN might hold some advantages compared to the other interfaces with respect to interaction and synchrony but studies on patients are needed to confirm these findings in the clinical setting.

Key messages

  • HN is a new interface that ameliorates its previous standard version, HS, by reducing the pressure dissipated because of the downward displacement of the soft collar during ventilator insufflation, and by avoiding the use of armpit braces.

  • We found that NIV via an FM, HS and HN at baseline settings, that is, PEEP 5 cmH2O and PS 5 cmH2O, were not significantly different with respect to inspiratory muscles unloading, WOB and comfort in healthy volunteers. With respect to patient-ventilator interaction and inspiratory trigger delay synchrony time and were significantly improved by FM, compared to both helmets, whereas expiratory trigger delay was also shorter with FM, as opposed to HS, but not HN.

  • At increased settings with HN (that is, PEEP 8 cmH2O and PS 8 cmH2O), inspiratory effort was significantly lower compared to that obtained while breathing spontaneously without assistance and FM. Patient-ventilator synchrony, as assessed by inspiratory trigger delay and expiratory trigger delay was equivalent between FM at baseline setting and HN at increased settings, but remained inferior with HS even at increased settings.

Abbreviations

Delay:

TR-exp: expiratory trigger delay

Delay:

TR-insp: inspiratory trigger delay

FiO2:

inspired oxygen fraction

FM:

face mask

HN:

helmet next

HR:

heart rate

HS:

helmet standard

NIV:

nonivasive ventilation

Pao:

pressure at the airway opening

Pdi:

transdiaphragmatic pressure

PEEP:

positive end-expiratory pressure

Pes:

esophageal pressure

Pga:

gastric pressure

PS:

pressure support

PSV:

pressure support ventilation

PTPdi:

pressure-time product of transdiaphragmatic pressure

RR:

respiratory rate

RRn:

patient’s neural respiratory rate

RRv:

ventilator rate of cycling

SB:

spontaneous breathing

SBE:

end spontaneous breathing

SBI:

initial spontaneous breathing

SpO2:

pulse arterial oxygen saturation

TI:

p: patient’s inspiratory time

TI:

synchrony: time of synchrony

TI:

v: ventilator’s inspiratory time

VAS:

visual analog scale

VE:

minute ventilation

VT:

tidal volume

WOB:

work of breathing.

References

  1. Kwok H, McCormack J, Cece R, Houtchens J, Hill NS: Controlled trial of oronasal versus nasal mask ventilation in the treatment of acute respiratory failure. Crit Care Med 2003, 31: 468-473. 10.1097/01.CCM.0000045563.64187.20

    Article  PubMed  Google Scholar 

  2. Navalesi P, Fanfulla F, Frigerio P, Gregoretti C, Nava S: Physiologic evaluation of noninvasive mechanical ventilation delivered with three types of masks in patients with chronic hypercapnic respiratory failure. Crit Care Med 2000, 28: 1785-1790. 10.1097/00003246-200006000-00015

    Article  CAS  PubMed  Google Scholar 

  3. Soo Hoo GW, Santiago S, Williams AJ: Nasal mechanical ventilation for hypercapnic respiratory failure in chronic obstructive pulmonary disease: determinants of success and failure. Crit Care Med 1994, 22: 1253-1261. 10.1097/00003246-199408000-00009

    Article  CAS  PubMed  Google Scholar 

  4. Gregoretti C, Confalonieri M, Navalesi P, Squadrone V, Frigerio P, Beltrame F, Carbone G, Conti G, Gamna F, Nava S, Calderini E, Skrobik Y, Antonelli M: Evaluation of patient skin breakdown and comfort with a new face mask for non-invasive ventilation: a multi-center study. Intensive Care Med 2002, 28: 278-284. 10.1007/s00134-002-1208-7

    Article  PubMed  Google Scholar 

  5. Meduri GU: Noninvasive positive-pressure ventilation in patients with acute respiratory failure. Clin Chest Med 1996, 17: 513-553. 10.1016/S0272-5231(05)70330-0

    Article  CAS  PubMed  Google Scholar 

  6. Kramer N, Meyer TJ, Meharg J, Cece RD, Hill NS: Randomized, prospective trial of noninvasive positive pressure ventilation in acute respiratory failure. Am J Respir Crit Care Med 1995, 151: 1799-1806. 10.1164/ajrccm.151.6.7767523

    Article  CAS  PubMed  Google Scholar 

  7. Antonelli M, Conti G, Pelosi P, Gregoretti C, Pennisi MA, Costa R, Severgnini P, Chiaranda M, Proietti R: New treatment of acute hypoxemic respiratory failure: noninvasive pressure support ventilation delivered by helmet–a pilot controlled trial. Crit Care Med 2002, 30: 602-608. 10.1097/00003246-200203000-00019

    Article  PubMed  Google Scholar 

  8. Antonelli M, Pennisi MA, Pelosi P, Gregoretti C, Squadrone V, Rocco M, Cecchini L, Chiumello D, Severgnini P, Proietti R, Navalesi P, Conti G: Noninvasive positive pressure ventilation using a helmet in patients with acute exacerbation of chronic obstructive pulmonary disease: a feasibility study. Anesthesiology 2004, 100: 16-24. 10.1097/00000542-200401000-00007

    Article  PubMed  Google Scholar 

  9. Chiumello D, Pelosi P, Carlesso E, Severgnini P, Aspesi M, Gamberoni C, Antonelli M, Conti G, Chiaranda M, Gattinoni L: Noninvasive positive pressure ventilation delivered by helmet vs. standard face mask. Intensive Care Med 2003, 29: 1671-1679. 10.1007/s00134-003-1825-9

    Article  PubMed  Google Scholar 

  10. Navalesi P, Costa R, Ceriana P, Carlucci A, Prinianakis G, Antonelli M, Conti G, Nava S: Non-invasive ventilation in chronic obstructive pulmonary disease patients: helmet versus facial mask. Intensive Care Med 2007, 33: 74-81. 10.1007/s00134-006-0391-3

    Article  PubMed  Google Scholar 

  11. Racca F, Appendini L, Gregoretti C, Stra E, Patessio A, Donner CF, Ranieri VM: Effectiveness of mask and helmet interfaces to deliver noninvasive ventilation in a human model of resistive breathing. J Appl Physiol 2005, 99: 1262-1271. 10.1152/japplphysiol.01363.2004

    Article  PubMed  Google Scholar 

  12. Vargas F, Thille A, Lyazidi A, Campo FR, Brochard L: Helmet with specific settings versus facemask for noninvasive ventilation. Crit Care Med 2009, 37: 1921-1928. 10.1097/CCM.0b013e31819fff93

    Article  PubMed  Google Scholar 

  13. Lucchini A, Valsecchi D, Elli S, Doni V, Corsaro P, Tundo P, Re R, Foti G, Manici M: The comfort of patients ventilated with the Helmet Bundle. Assist Inferm Ric 2010, 29: 174-183.

    PubMed  Google Scholar 

  14. Olivieri C, Costa R, Spinazzola G, Ferrone G, Longhini F, Cammarota G, Conti G, Navalesi P: Bench comparative evaluation of a new generation and standard helmet for delivering non-invasive ventilation. Intensive Care Med 2013, 39: 734-738. 10.1007/s00134-012-2765-z

    Article  CAS  PubMed  Google Scholar 

  15. Vaschetto R, Turucz E, Dellapiazza F, Guido S, Colombo D, Cammarota G, Della CF, Antonelli M, Navalesi P: Noninvasive ventilation after early extubation in patients recovering from hypoxemic acute respiratory failure: a single-centre feasibility study. Intensive Care Med 2012, 38: 1599-1606. 10.1007/s00134-012-2652-7

    Article  PubMed  Google Scholar 

  16. Girault C, Briel A, Benichou J, Hellot MF, Dachraoui F, Tamion F, Bonmarchand G: Interface strategy during noninvasive positive pressure ventilation for hypercapnic acute respiratory failure. Crit Care Med 2009, 37: 124-131. 10.1097/CCM.0b013e3181928706

    Article  PubMed  Google Scholar 

  17. Crimi C, Noto A, Princi P, Esquinas A, Nava S: A European survey of noninvasive ventilation practices. Eur Respir J 2010, 36: 362-369. 10.1183/09031936.00123509

    Article  CAS  PubMed  Google Scholar 

  18. Baydur A, Behrakis PK, Zin WA, Jaeger M, Milic-Emili J: A simple method for assessing the validity of the esophageal balloon technique. Am Rev Respir Dis 1982, 126: 788-791.

    CAS  PubMed  Google Scholar 

  19. Appendini L, Purro A, Gudjonsdottir M, Baderna P, Patessio A, Zanaboni S, Donner CF, Rossi A: Physiologic response of ventilator-dependent patients with chronic obstructive pulmonary disease to proportional assist ventilation and continuous positive airway pressure. Am J Respir Crit Care Med 1999, 159: 1510-1517. 10.1164/ajrccm.159.5.9804130

    Article  CAS  PubMed  Google Scholar 

  20. Barnard PA, Levine S: Critique on application of diaphragmatic time-tension index to spontaneously breathing humans. J Appl Physiol 1986, 60: 1067-1072.

    CAS  PubMed  Google Scholar 

  21. Costa R, Navalesi P, Antonelli M, Cavaliere F, Craba A, Proietti R, Conti G: Physiologic evaluation of different levels of assistance during noninvasive ventilation delivered through a helmet. Chest 2005, 128: 2984-2990. 10.1378/chest.128.4.2984

    Article  PubMed  Google Scholar 

  22. Brochard L, Rua F, Lorino H, Lemaire F, Harf A: Inspiratory pressure support compensates for the additional work of breathing caused by the endotracheal tube. Anesthesiology 1991, 75: 739-745. 10.1097/00000542-199111000-00004

    Article  CAS  PubMed  Google Scholar 

  23. Fraticelli AT, Lellouche F, L’her E, Taille S, Mancebo J, Brochard L: Physiological effects of different interfaces during noninvasive ventilation for acute respiratory failure. Crit Care Med 2009, 37: 939-945. 10.1097/CCM.0b013e31819b575f

    Article  PubMed  Google Scholar 

  24. Jaber S, Fodil R, Carlucci A, Boussarsar M, Pigeot J, Lemaire F, Harf A, Lofaso F, Isabey D, Brochard L: Noninvasive ventilation with helium-oxygen in acute exacerbations of chronic obstructive pulmonary disease. Am J Respir Crit Care Med 2000, 161: 1191-1200. 10.1164/ajrccm.161.4.9904065

    Article  CAS  PubMed  Google Scholar 

  25. Gift AG: Visual analogue scales: measurement of subjective phenomena. Nurs Res 1989, 38: 286-288.

    Article  CAS  PubMed  Google Scholar 

  26. Vitacca M, Bianchi L, Zanotti E, Vianello A, Barbano L, Porta R, Clini E: Assessment of physiologic variables and subjective comfort under different levels of pressure support ventilation. Chest 2004, 126: 851-859. 10.1378/chest.126.3.851

    Article  PubMed  Google Scholar 

  27. Perez W, Tobin MJ: Separation of factors responsible for change in breathing pattern induced by instrumentation. J Appl Physiol 1985, 59: 1515-1520.

    CAS  PubMed  Google Scholar 

  28. Weissman C, Askanazi J, Milic-Emili J, Kinney JM: Effect of respiratory apparatus on respiration. J Appl Physiol 1984, 57: 475-480.

    CAS  PubMed  Google Scholar 

Download references

Acknowledgements

This study was supported by University Hospital of Montpellier.

Author information

Authors and Affiliations

Authors

Corresponding author

Correspondence to Samir Jaber.

Additional information

Competing interests

StarMed financially contributed by providing helmets (NIV-Castar R Next and NIV-Castar R) and double balloon-tipped catheters. PN contributed to the development of the helmet NIV-Castar R Next, for which the licence for the patent belongs to Starmed SPA, and receives royalties for that invention. SJ was a paid consultant to Starmed SPA. RV, ADJ, MC, FG, MM, YC, AP have no conflicts of interest to declare.

Authors’ contributions

RV contributed to the conception and design of the study, to the analysis and interpretation of data, to drafting the submitted article, and to providing final approval of the version to be published. ADJ contributed to the analysis of the data, and to providing final approval of the version to be published. MC contributed to the acquisition of the data, and to providing final approval of the version to be published. FG contributed to the acquisition of the data, and to providing final approval of the version to be published. MM contributed to the acquisition of the data, and to providing final approval of the version to be published. YC contributed to the acquisition of the data, and to providing final approval of the version to be published. PA contributed to the acquisition of the data, and to providing final approval of the version to be published. PN contributed to the conception of the study, interpretation of data and drafting of the submitted article, and to providing final approval of the version to be published. SJ contributed to the conception and design of the study; to the analysis and interpretation of data; to drafting the submitted article, and to providing final approval of the version to be published. All authors read and approved the final manuscript.

Electronic supplementary material

13054_2012_2190_MOESM1_ESM.docx

Additional file 1: Tables E1 and E2: An additional file in the appendix contains two additional tables: Table E1 shows the main ventilatory variables and inspiratory effort parameters obtained at baseline settings and following interventions at increased settings, obtained in spontaneous breathing (SB), facial mask (FM), standard helmet (HS) and helmet next (HN) with data presented as median [interquartiles]; and Table E2 shows the inspiratory and expiratory trigger delay and synchrony time obtained at baseline settings and following interventions at increased settings, obtained in facial mask (FM), standard helmet (HS) and helmet next (HN) with data presented as median [interquartiles]. (DOCX 29 KB)

Authors’ original submitted files for images

Rights and permissions

Reprints and permissions

About this article

Cite this article

Vaschetto, R., De Jong, A., Conseil, M. et al. Comparative evaluation of three interfaces for non-invasive ventilation: a randomized cross-over design physiologic study on healthy volunteers. Crit Care 18, R2 (2014). https://doi.org/10.1186/cc13175

Download citation

  • Received:

  • Accepted:

  • Published:

  • DOI: https://doi.org/10.1186/cc13175

Keywords