Potential bias and misclassification of using continuous cardiac output to identify fluid responsiveness compared to calibrated measurements

Introduction

Fluid responsiveness was cornered as being of preeminent importance to optimize hemodynamics during circulatory shock [1]. This is facilitated in clinical routine by continuous cardiac output (CCO) monitoring. Yet, a theoretical risk exists of misclassification of fluid responsiveness if one uses the CCO value measured at the end of a fluid challenge (FC) without recalibrating the device. We hence evaluated the bias existing between calibrated cardiac output (CO) measured at FC’s end and the value of CCO measured immediately before that same re-calibration.

Materials and methods

We report an ancillary of an observational single-center study performed in a tertiary ICU in Lyon, France. The study was approved by an ethics comity (Comité Scientifique et Ethique des Hospices Civils de Lyon, reference 23–5040). We enrolled consecutive patients with circulatory shock receiving norepinephrine and calibrated CCO monitoring (PiCCO®, Pulsion Medical, Germany), and who received a 500-ml FC of crystalloids in less than 15 min.

The primary outcome was the bias between the recalibrated CO (method 1) and the CCO value measured immediately at FC’s end and prior to the device recalibration (method 2). Secondary outcomes evaluated the trending and diagnostic performance of method 2 to identify fluid responsiveness.

The calibration was performed by mean of transpulmonary thermodilution (TPTD, 3 × injections of 15-ml cold saline). The device was calibrated twice, immediately before (T1) and immediately at FC’s end (T2, recalibration) to obtain calibrated CO (COTPTD). CCO by pulse contour analysis was collected twice (continuous recording of the 12-s moving average of beat-to-beat CCO refreshed and sampled at 1 Hz): immediately after T1 (mean value over a 60-s stable hemodynamic period) and immediately before T2 (mean value over the last 60 s). We computed the respective relative changes in CCO (∆%CCO) and COTPTD (∆%COTPTD) between T1 and T2. Fluid responsiveness was adjudicated if ∆%COTPTD increased > 15%.

Data were reported by their median [interquartile range], mean ± standard deviation, or count (percentage). Bias was evaluated using a Bland–Altman representation. Ability of ∆%CCO to track changes in ∆%COTPTD was assessed using 4-quadrant and radial plots. The diagnostic performance of ∆%CCO to identify fluid responsiveness was assessed using the area under the receiver operating curve (AUROC). 95% confidence intervals (95%c.i.) were computed using bootstrapping (n = 1000).

Results

Between April 4th, 2023 and June 16th, 2023, we enrolled 15 patients within a delay of 1 [0.5–1] day after ICU admission (Supplemental Table 1 for the population characteristics). The elapsed time between T1 and T2 was 18 [14–20] min, and FCs were administered over 7 [7–8] min.

The bias between methods at T2 was − 0.29 ± 0.70 L.min−1 (constant bias, limits of agreements ± 1.4 L.min−1). The CCO method had a percentage error of 25% (95%c.i.: 14%–36%) against COTPTD (Fig. 1A). ∆%CCO demonstrated intermediate trending ability to detect changes in ∆%COTPTD (Fig. 1B), with a concordance rate of 79% (95%c.i.: 49%–94%). The radial plot showed an angular bias between methods of − 3° ± 35° (Fig. 1C).

Fig. 1
figure 1

The figure shows the bias between CCO and COTPTD measured at the end of a fluid challenge (panel A), the concordance between the relative change in CCO (∆%CCO) and COTPTD (∆%COTPTD) during the fluid challenge (panel B), the radial plot quantifying the bias in relative change (panel C) and the difference in relative change in CCO during the fluid challenge in patients identified as being fluid responders or non-responders (classified using a ∆%COTPTD threshold > 15%, panel D). Panel A is a Bland and Altman plot showing a constant bias (mean bias − 0.29 ± 0.70 L min−1, limits of agreements ± 1.4 L min−1). Panel B is a concordance plot, with concordant measurements situated in the north-eastern and south-western quadrants. Panel C is a radial plot showing the angular bias between ∆%CCO and ∆%COTPTD, identified by the broad line, and the radial limits of agreement (dashed lines). The angular bias (− 3° ± 35°) was statistically different from 0° (P = 0.39), with radial limits of agreements of ± 71°. CCO: continuous cardiac output; COTPTD: calibrated cardiac output by transpulmonary thermodilution; ∆%CCO: relative change in CCO between T1 and T2 (before the second calibration); ∆%COTPTD: relative change in COTPTD between T1 et T2

Fluid responsiveness was identified in 7/15 of FCs using COTPTD. ∆%CCO was significantly higher in fluid responders compared to non-responders (Fig. 1D), although ∆%CCO was < 15% in 2 fluid-responsive cases. ∆%CCO had an AUROC of 0.83 (95%c.i.: 0.60–1.00, P = 0.04) to identify fluid responsiveness. At the threshold of 15%, ∆%CCO had a sensitivity of 0.70 (95%c.i.: 0.38–1.00) and a specificity > 0.99 (95%c.i.: 0.99–1.00) to identify fluid responsiveness.

Discussion and conclusion

In this single-center observational study, we identified that 1/ CCO measured immediately before CO recalibration after a FC demonstrated a small negative bias; 2/ ∆%CCO demonstrated intermediate trending capacity with potentially large bias between methods; and 3/ ∆%CCO had acceptable classifying performance to identify fluid responsiveness, with a risk of false negative results.

Our findings suggest that, while performing a FC monitored by calibrated CCO, cautious interpretation of the FC’s results should be made, due to potential bias impacting its relative change from baseline. The pharmacokinetics of a FC show that the infusion of 500 ml of crystalloid at 20 °C may not only improve venous return and potentially CO, but could also alter arterial or venous compliance and resistance [2]. These modifications will eventually modify the arterial root signal of CCO, and lead to misclassification [3].

FC’s hemodynamic effect dissipation occurring between the end of the FC and the end of recalibration (5 min) may not be retained, as the bias between method was negative (i.e. CCO was lower than COTPTD), and no cases showed a ∆%CCO > 15% in non-responders [4]. Finally, COTPTD measured by triplicate injection demonstrates a precision of 7% and least significant change (LSC) of 10%, which implies potentially inaccurate adjudication of fluid responsiveness using this technique [5].

To conclude, using CCO to evaluate fluid responsiveness in patients receiving a FC has the advantage of being efficient, but goes with the risk of misclassification and misleading clinical conclusions.

Availability of data and materials

Source datasets are not publicly available due to ethical reasons. Further enquiries can be directed to the corresponding author.

Abbreviations

AUROC:

Area under the receiver-operating curve

CCO:

Continuous cardiac output

COTPTD :

Calibrated cardiac output by transpulmonary thermodilution

∆%CCO:

Relative change in CCO between T1 and T2 (before the second calibration)

∆%COTPTD :

Relative change in COTPTD between T1 et T2

FC:

Fluid challenge

TPTD:

Transpulmonary thermodilution

References

  1. Cecconi M, De Backer D, Antonelli M, Beale R, Bakker J, Hofer C, Jaeschke R, Mebazaa A, Pinsky MR, Teboul JL, et al. Consensus on circulatory shock and hemodynamic monitoring task force of the european society of intensive care medicine. Intensive Care Med. 2014;40(12):1795–815.

    Article  PubMed  PubMed Central  Google Scholar 

  2. Bitker L, Cutuli SL, Yanase F, Wilson A, Osawa EA, Lucchetta L, Cioccari L, Canet E, Glassford N, Eastwood GM, et al. The hemodynamic effects of warm versus room-temperature crystalloid fluid bolus therapy in post-cardiac surgery patients. Perfusion. 2022;37:613–23.

    Article  PubMed  Google Scholar 

  3. Hamzaoui O, Monnet X, Richard C, Osman D, Chemla D, Teboul JL. Effects of changes in vascular tone on the agreement between pulse contour and transpulmonary thermodilution cardiac output measurements within an up to 6-hour calibration-free period. Crit Care Med. 2008;36(2):434–40.

    Article  PubMed  Google Scholar 

  4. Aya HD, Ster IC, Fletcher N, Grounds RM, Rhodes A, Cecconi M. Pharmacodynamic analysis of a fluid challenge. Crit Care Med. 2016;44(5):880–91.

    Article  CAS  PubMed  Google Scholar 

  5. Giraud R, Siegenthaler N, Merlani P, Bendjelid K. Reproducibility of transpulmonary thermodilution cardiac output measurements in clinical practice: a systematic review. J Clin Monit Comput. 2017;31(1):43–51.

    Article  PubMed  Google Scholar 

Download references

Acknowledgements

The authors are grateful to all patients and their families who participated to the study.

Funding

None.

Author information

Authors and Affiliations

Authors

Contributions

LB designed the study, collected and analyzed the data, interpreted the results, and drafted the manuscript. GD and JCR interpreted the results and revised the manuscript for important intellectual content. All authors read and approved the manuscript submitted for publication.

Corresponding author

Correspondence to Laurent Bitker.

Ethics declarations

Ethics approval and consent to participate

The study was conducted in accordance with the Declaration of Helsinki and with local regulations. The study obtained ethics approval from the Comité Scientifique et Ethique des Hospices Civils de Lyon, under the reference number 23–5040. All participants or their next-of-kin received all required information regarding study procedures. Due to the non-interventional design of the study, the ethics comity waived the obligation for signed consent.

Consent for publication

Not applicable.

Competing interests

The authors declare that they have no competing interest.

Additional information

Publisher's Note

Springer Nature remains neutral with regard to jurisdictional claims in published maps and institutional affiliations.

Supplementary Information

Rights and permissions

Open Access This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.

Reprints and permissions

About this article

Check for updates. Verify currency and authenticity via CrossMark

Cite this article

Bitker, L., Deniel, G. & Richard, JC. Potential bias and misclassification of using continuous cardiac output to identify fluid responsiveness compared to calibrated measurements. Crit Care 28, 202 (2024). https://doi.org/10.1186/s13054-024-04993-1

Download citation

  • Received:

  • Accepted:

  • Published:

  • DOI: https://doi.org/10.1186/s13054-024-04993-1