Meta-analysis of pulse pressure variation (PPV) and stroke volume variation (SVV) studies: a few rotten apples can spoil the whole barrel

Matters Arising to this article was published on 12 January 2024

The Original Article was published on 08 November 2023

We read the meta-analysis by Messina et al. [1] on the effectiveness of pulse pressure variation (PPV) and stroke volume variation (SVV) in predicting fluid responsiveness with both interest and skepticism.

This meta-analysis included studies involving patients with an open chest, ventilated with a low tidal volume, or undergoing laparoscopic surgery (with increased abdominal pressure due to pneumoperitoneum). In these conditions, PPV and SVV are known to be unreliable to predict fluid responsiveness [2]. Including studies where these limitations are not respected inevitably leads to an overall moderate predictive value. In other words, it is entirely foreseeable that the performance of a diagnostic tool will be moderate when one fails to consider the well-known limitations associated with its use.

We appreciate the fact that the predictive value of PPV and SVV was assessed in various subgroups. Regrettably, it was not evaluated in the subgroup of patients meeting all the conditions conducive to the reliable use of PPV and SVV. What was the area under the curve (AUC) in the subgroup of patients undergoing non-laparoscopic surgery with a closed chest and a tidal volume of 7–9 ml/kg? Would it support the conclusion that PPV and SVV are only moderately accurate to predict fluid responsiveness?

We firmly believe that both PPV and SVV serve as reliable predictors of fluid responsiveness, provided that physiologic limitations to their use are respected [3, 4]. Since the initial description of PPV almost 25 yrs ago [5], these limitations have been extensively discussed in numerous articles, including in this journal [2], and are summarized once more in the “PPV-meter” shown in Fig. 1. Many of these limitations (e.g. atrial fibrillation, spontaneous breathing activity, low tidal volume) are encountered less frequently in patients undergoing major surgery with general anesthesia than in critically ill patients. Of note, a tidal volume of 7–9 ml/kg, commonly used during surgery [6], has been deemed safe [7]. For patients ventilated with a tidal volume < 7 ml/kg, the assessment of changes in PPV during a mini-fluid challenge [8] or a transient rise in tidal volume (aka tidal volume challenge) [9] has proved useful to predict fluid responsiveness. Unfortunately, these points were not addressed in the paper by Messina et al. [1].

Fig. 1
figure 1

Pulse pressure variation (PPV)-meter summarizing the clinical meaning of PPV (right) and main limitations to its clinical use (left)

Finally, unlike SVV monitoring, PPV monitoring does not require any cardiac output monitoring device. In a meta-analysis assessing the respective performance of PPV and SVV, it would have been wise to highlight this practical advantage as well.

In summary, physiologic limitations to the use of PPV and SVV should be respected not only in clinical practice but also in meta-analyses; otherwise, they may lead to misleading conclusions. When these limitations are respected, we believe that both PPV and SVV are valuable variables for predicting fluid responsiveness and personalizing hemodynamic management, potentially leading to improved patient outcomes [10].

Availability of data and material

Not applicable.

Abbreviations

PPV:

Pulse pressure variation

SVV:

Stroke volume variation

AUC:

Area under the curve

References

  1. Messina A, Caporale M, Calabro L, et al. Reliability of pulse pressure and stroke volume variation in assessing fluid responsiveness in the operating room: a metaanalysis and a metaregression. Crit Care. 2023;27:431.

    Article  PubMed  PubMed Central  Google Scholar 

  2. Michard F, Chemla D, Teboul JL. Applicability of pulse pressure variation: how many shades of grey? Crit Care. 2015;19:144.

    Article  PubMed  PubMed Central  Google Scholar 

  3. Michard F. Changes in arterial pressure during mechanical ventilation. Anesthesiology. 2005;103:419–28.

    Article  PubMed  Google Scholar 

  4. Teboul JL, Monnet X, Chemla D, Michard F. Arterial pulse pressure variation with mechanical ventilation. Am J Respir Crit Care Med. 2019;199:22–31.

    Article  PubMed  Google Scholar 

  5. Michard F, Chemla D, Richard C, et al. Clinical use of respiratory changes in arterial pulse pressure to monitor the hemodynamic effects of PEEP. Am J Respir Crit Care Med. 1999;159:935–9.

    Article  CAS  PubMed  Google Scholar 

  6. LAS Vegas investigators. Epidemiology, practice of ventilation and outcome for patients at increased risk of postoperative pulmonary complications: LAS VEGAS – an observational study in 29 countries. Eur J Anaesthesiol. 2017;34:492–507.

    Article  Google Scholar 

  7. Levin MA, McCormick PJ, Lin HM, et al. Low intraoperative tidal volume ventilation with minimal PEEP is associated with increased mortality. Br J Anaesth. 2014;113:97–108.

    Article  CAS  PubMed  Google Scholar 

  8. Mallat J, Meddour M, Durville E, et al. Decrease in pulse pressure and stroke volume variation after mini-fluid challenge accurately predicts fluid responsiveness. Br J Anaesth. 2015;115:449–56.

    Article  CAS  PubMed  Google Scholar 

  9. Myatra SN, Monnet X, Teboul JL. Use of tidal volume challenge to improve the reliability of pulse pressure variation. Crit Care. 2017;21:60.

    Article  PubMed  PubMed Central  Google Scholar 

  10. Benes J, Giglio M, Brienza N, Michard F. The effects of goal-directed fluid therapy based on dynamic parameters on post-surgical outcome: a meta-analysis of randomized controlled trials. Crit Care. 2014;18:584.

    Article  PubMed  PubMed Central  Google Scholar 

Download references

Funding

The authors did not receive any funding for this article.

Author information

Authors and Affiliations

Authors

Contributions

FM drafted the manuscript, and all authors revised the manuscript and approved the final version.

Corresponding author

Correspondence to Frederic Michard.

Ethics declarations

Ethics approval and consent to participate

Not applicable.

Competing interests

None in relation with this article.

Additional information

Publisher's Note

Springer Nature remains neutral with regard to jurisdictional claims in published maps and institutional affiliations.

This comment refers to the article available online at https://doi.org/10.1186/s13054-023-04706-0.

Rights and permissions

Open Access This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.

Reprints and permissions

About this article

Check for updates. Verify currency and authenticity via CrossMark

Cite this article

Michard, F., Chemla, D. & Teboul, JL. Meta-analysis of pulse pressure variation (PPV) and stroke volume variation (SVV) studies: a few rotten apples can spoil the whole barrel. Crit Care 27, 482 (2023). https://doi.org/10.1186/s13054-023-04765-3

Download citation

  • Received:

  • Accepted:

  • Published:

  • DOI: https://doi.org/10.1186/s13054-023-04765-3

Keywords