Circuit change during extracorporeal membrane oxygenation: single-center retrospective study of 48 changes

Abstract

Background

Bleeding and thrombosis induce major morbidity and mortality in patients under extracorporeal membrane oxygenator (ECMO). Circuit changes can be performed for oxygenation membrane thrombosis but are not recommended for bleeding under ECMO. The objective of this study was to evaluate the course of clinical, laboratory, and transfusion parameters before and after ECMO circuit changes warranted by bleeding or thrombosis.

Methods

In this single-center, retrospective, cohort study, clinical parameters (bleeding syndrome, hemostatic procedures, oxygenation parameters, transfusion) and laboratory parameters (platelet count, hemoglobin, fibrinogen, PaO2) were collected over the seven days surrounding the circuit change.

Results

In the 274 patients on ECMO from January 2017 to August 2020, 48 circuit changes were performed in 44 patients, including 32 for bleeding and 16 for thrombosis. Mortality was similar in the patients with vs. without changes (21/44, 48% vs. 100/230, 43%) and in those with bleeding vs. thrombosis (12/28, 43% vs. 9/16, 56%, P = 0.39). In patients with bleeding, numbers of bleeding events, hemostatic procedures, and red blood cell transfusions were significantly higher before vs. after the change (P < 0.001); the platelet counts and fibrinogen levels decreased progressively before and increased significantly after the change. In patients with thrombosis, numbers of bleeding events and red blood cell transfusions did not change after membrane change. No significant differences were demonstrated between oxygenation parameters (ventilator FiO2, ECMO FiO2, and PaO2) and ECMO flow before vs. after the change.

Conclusions

In patients with severe and persistent bleeding, changing the ECMO circuit decreased clinical bleeding and red blood cell transfusion needs and increased platelets and fibrinogen levels. Oxygenation parameters did not change significantly in the group with thrombosis.

Background

Extracorporeal membrane oxygenation (ECMO) is a recognized treatment for refractory hypoxemia (veno-venous ECMO, VV-ECMO) and cardiogenic shock (veno-arterial ECMO, VA-ECMO) [1]. Bleeding is common during ECMO (34% of patients on VV-ECMO and 44% on VA-ECMO), whereas thrombosis is rarer (11% on VV-ECMO and 7% on VA-ECMO) [2]. Bleeding and thrombosis induce major morbidity and mortality [2], and their prevention is a major treatment goal. Recent guidelines by the Extracorporeal Life Support Organization (ELSO) state that anticoagulation should be used for most patients whose anti-factor Xa (antiXa) level is within the 0.3–0.7 IU/mL range [3]. Thrombosis can require replacement of the membrane oxygenator [3]. Although performed in clinical practice, changing the membrane is not mentioned in any guidelines as part of the management of bleeding [4].

The objective of this study was to evaluate the course of clinical, laboratory, and transfusion parameters after ECMO circuit changes warranted by bleeding or thrombosis.

Methods

This retrospective study included patients managed in the 23-bed adult intensive care unit of the Marie Lannelongue Cardiothoracic Surgery Hospital (Groupe Hospitalier Paris Saint Joseph, Le Plessis-Robinson, France). A national review board CERAR approved the study (IRB0001025). All patients received an information and no-objection statements in accordance with the French law on retrospective studies.

Records of consecutive adults managed with ECMO between January 2017 and August 2020 were screened to identify patients who required circuit changes due to bleeding or thrombosis with oxygenation failure. Each circuit change was handled as a separate event.

Definitions

Bleeding group was defined as either bleeding persistent for more than two days or bleeding uncontrolled despite hemostatic procedures (arterio-embolization, surgical revision, packing, local hemostatic procedure during endoscopy). Bleeding localization was epistaxis or oral bleeding, cannula- or catheter-related bleeding, hemoptysis, and deep bleeding.

Thrombosis group included sudden thrombosis of head pump or circuit, clots on the membrane visible to the naked eye combined with hypoxemia defined as desaturation < 90% or right radial PaO2 < 80 mmHg with ECMO FiO2 ≥ 80%. When oxygenator did not allow direct visualization of clots on the membrane, thrombosis was defined as membrane dysfunction with a post-oxygenator blood gas sample showing an ECMO PaO2/FiO2 < 200 mmHg [5].

ECMO monitoring

For VV- and VA-ECMO, unfractionated heparin (UFH) was administered to maintain antiXa between 0.2 and 0.4 IU/mL, with dosage adjustments after each antiXa assay performed at least every 12 h. In compliance with ELSO recommendations, in patients with bleeding, UFH was interrupted for up to 12 h or as long as needed to control the bleeding [3]. Once the bleeding was controlled, effective anticoagulation was reintroduced in patients with formal indications (e.g., pulmonary embolism, loss of arterial pulsatility, mechanical valve).

Laboratory tests performed at least twice daily included hemoglobin, platelet count, antiXa, fibrinogen, pH, PaCO2, and PaO2.

Protocolized transfusions of red blood cells (RBC), platelets, fresh frozen plasma (FFP), and fibrinogen were given to maintain hemoglobin > 7 g/dL, platelets > 30 000/mm3, fibrinogen > 1.5 g/dL, and prothrombin time > 50% [6].

Data collection

For this study, we collected data over the seven days before and after the circuit change (from D-7 to D + 7). D-1 and D + 1 corresponded to the 24 h before and after the change. The data included the blood tests listed above, transfusion data (administered amounts of RBC units, FFP units, platelets, and fibrinogen), and technical data (ECMO and ventilator FiO2, ECMO flow rate). Clinical bleeding and hemostatic procedures were collected over the five days before and after the change (from D-5 to D + 5); this shorter collection period limited the risk of missing data.

Statistical analysis

Data were analyzed using StatView 5.0 software (SAS Institute, Cary, NC). Data distribution was assessed by the Kolmogorov–Smirnov test. Normal data were described as mean ± SD and non-normal data as median [interquartile range]. Continuous variables were compared using Student's t test; for repeated measures, we performed ANOVA followed by the Scheffe test. P values < 0.05 were considered significant.

Results

Patients

We identified 274 patients managed with VA-ECMO (n = 211, 77%) or VV-ECMO (n = 63, 23%) during the study period. The proportion of patients with circuit changes was 18% (49 patients, 53 changes) overall and 31% (46 patients) among the 147 patients on ECMO for longer than seven days. We excluded five patients who were transferred on ECMO from another hospital, which wants to get back the ECMO device. For these patients, a scheduled change was performed. The remaining 44 patients had 48 circuit changes, including 32 (67%) for bleeding and 16 (33%) for thrombosis; two patients with bleeding had three changes each. The table reports the main patient characteristics and outcomes. Mortality was similar in the patients with vs. without circuit changes (21/44 (48%) vs. 100/230 (43%) P = 0.60) and in the patients with bleeding vs. thrombosis (Table 1).

Table 1 Characteristics and outcomes of the 44 patients with extracorporeal membrane oxygenation circuit changes

Group with bleeding (n = 32)

Compared to the five days before the circuit change, the five days after the change had significantly fewer clinical bleeding events (29 vs. 84, P < 0.001) and hemostatic procedures (4 vs. 19, P = 0.001) (Fig. 1A). Significant rises in fibrinogen (P < 0.001) and platelet (P < 0.001) levels occurred after the change (Fig. 1B), although the amounts of fibrinogen and platelets administered were not significantly different before and after the change (P = 0.31 and P = 0.11, respectively).

Fig. 1
figure 1

Main results in the group with bleeding (panels A and B) and in the group with thrombosis (panel C and D). A Group with bleeding: clinical bleeding events and hemostatic procedures during the five days before and the five days after the circuit change. The X-axis shows time in days; the black dotted line indicates the time of the circuit change. The Y-axis on the left shows the number of clinical bleeding events (gray bars) and the Y-axis on the right the number of hemostatic procedures (white bars). During the 5 days before the membrane change, 84 bleeding events occurred, 19 of which required a hemostatic procedure. In the 5 days following the membrane change, 29 bleeding events occurred, 4 of which required a hemostatic procedure. B Group with bleeding: changes in platelet counts and fibrinogen levels during the seven days before and the seven days after the circuit change. The X-axis shows time in days; the black dotted line indicates the time of the circuit change. The Y-axis on the left shows the platelet counts in G/L (gray bars) and the Y-axis on the right the fibrinogen levels in g/L (white bars). The ANOVA showed a significant difference in platelet (F = 5.7 P < 0.001) and fibrinogen values (F = 3.7 P < 0.001). In the period before the change (D-7 to D-1), the platelet values differed significantly (F = 10.2 P < 0.001). The decrease began at D-4 for a value of 34G/L P < 0.01. In the period after the membrane change (D + 1 to D + 7), the platelet values differed significantly (F = 7.2 P < 0.001). The increase started at D + 6 for a value of 73 G/L P = 0.01. In the period before the change (D-7 to D-1), the fibrinogen level differed significantly (F = 10.0 P < 0.001). The decrease began at D-3 for a value of 1.3 g/L P = 0.01. In the period after the membrane change (D + 1 to D + 7), the fibrinogen level differed significantly (F = 3.6 P = 0.002). The increase started at D + 5 for a value of 1.2 g/L P = 0.04. C Group with thrombosis: clinical bleeding events and hemostatic procedures during the five days before and the five days after the circuit change. The X-axis shows time in days; the black dotted line indicates the time of the circuit change. The Y-axis on the left shows the number of clinical bleeding events (gray bars) and the Y-axis on the right the number of hemostatic procedures (white bars). During the 5 days before the membrane change, 22 bleeding events occurred, 6 of which required a hemostatic procedure. In the 5 days following the membrane change, 24 bleeding events occurred, 7 of which required a hemostatic procedure. D Group with thrombosis: changes in platelet counts and fibrinogen levels during the seven days before and the seven days after the circuit change. The X-axis shows time in days; the black dotted line indicates the time of the circuit change. The Y-axis on the left shows the platelet counts in G/L (gray bars) and the Y-axis on the right the fibrinogen levels in g/L (white bars). The ANOVA showed a significant difference in platelet values (F = 2.0 P = 0.03). The ANOVA did not show a significant difference in fibrinogen level (F = 0.7 P = 0.71). In the period before the change (D-7 to D-1), the platelet values differed significantly (F = 2.9 P = 0.01). Individual comparison of platelet level showed no statistical difference. In the period after the membrane change (D + 1 to D + 7), the platelet values differed significantly (F = 2.8 P = 0.02). Individual comparison of platelet level showed no statistical difference. Before or after the membrane change, the fibrinogen values did not vary significantly (P = 0.12) Individual comparison of fibrinogen levels did not show any statistical difference

Compared to the pre-change period, the post-change period showed significant decreases in the numbers of administered RBC units (from 7.3 ± 6.4 to 3.7 ± 3.6, P = 0.01) and FFP units (from 3.0 ± 6.4 to 0.5 ± 1.4, P = 0.04). AntiXa level decreased after the change (from 0.20 ± 0.12 to 0.18 ± 0.10, P = 0.02).

Oxygenation did not improve after the change: ECMO FiO2, ventilator FiO2, PaO2, and ECMO flow and rotations per minute did not change significantly (data not shown).

Group with thrombosis (n = 16)

Clinical bleeding events (22 vs. 24, P = 0.76) and hemostatic procedures (6 vs. 7, P = 0.78) did not differ before and after membrane change (Fig. 1C).

Comparing before and after the change, platelet level varied significantly (P = 0.03), while the fibrinogen level remained stable (P = 0.71) (Fig. 1D). Platelets and fibrinogen administered were not significantly different before and after the change (P = 0.65 and P = 0.70, respectively).

Compared with the pre-change period, the post-change period showed no significant decreases in the numbers of administered RBC units (from 3.2 ± 1.9 to 2.3 ± 1.7, P = 0.12) and FFP units (from 0.6 ± 1.4 to 0.2 ± 0.7, P = 0.37).

None of the oxygenation parameters improved significantly after the circuit change (ECMO FiO2, P = 0.82; ventilator FiO2, P = 0.34; and PaO2, P = 0.21).

Discussion

The findings from this single-center retrospective study suggest that ECMO circuit changes in patients with persistent bleeding decrease the frequency of clinical bleeding events, the need for hemostatic procedures, and RBC and FFP transfusions. In contrast, in patients with thrombosis, such changes do not reduce the number of clinical bleeding events and hemostatic procedures or RBC and FFP transfusion. Oxygenation parameters did not improve in both groups after change. It is worth noting that our study was performed in a cardiothoracic surgery center and included over 70% of surgical patients mostly on VA-ECMO, whereas previous studies included only patients on VV-ECMO to treat acute respiratory distress syndrome.

Bleeding during ECMO is caused by several mechanisms that are dependent on ECMO duration. Over time, ECMO is associated with reductions in RBC deformability and with RBC clumping [7]. Blood in the circuit is subjected to high shear stress, which can cause hemolysis [8]. Finally, the increased consumption of fibrinogen and platelets also promotes bleeding [9]. The mechanisms of membrane oxygenator dysfunction may be interrelated, with thrombosis inducing hemolysis, which may in turn promote bleeding. The significant change in platelet levels in patients (thrombosis and bleeding) may be related to the physical characteristics of the circuit that play a role in platelet activation. Adherent platelets release their granules that contain proinflammatory cytokines and hemostatic factors [10]. Device-induced coagulation disorder is common, accounting for about 40% of circuit changes, but is accompanied with clinical bleeding in only half the cases [5].

The management of bleeding during ECMO is poorly standardized, with a consensus existing only for temporary withholding of anticoagulation [3]. The largest published cohort study (83 patients with at least one change) did not assess whether withholding anticoagulation decreased clinical bleeding [5]. Elevations in plasma-free hemoglobin (pfHb), serum lactate dehydrogenase (LDH), and D-dimer and declines in fibrinogen levels, platelet counts, and heparin doses may predict a need for a circuit change [5, 8]. However, D-dimer, pfHb, and LDH were not routinely measured in our center.

In the few patients with membrane thrombosis, hypoxemia persisted even after the membrane change. The lack of oxygenation improvement may have been related to a deterioration in patients lung function. Thus, the oxygenation might have worsened further, even if the circuits had not been change.

The main limitations of this study are the retrospective design and single-center recruitment of patients who underwent highly specific surgical procedures. Another limitation in the group with bleeding may be the exposure to heparin leached from the circuit over the weeks before the change but the day of the circuit change, the anticoagulation level remained low (median antiXa 0.08 IU/L [0–0.22]) despite some patients had formal indications for effective anticoagulation. The difference between pre- and post-membrane pressures was not collected, and we were therefore unable to accurately time the onset of membrane thrombosis.

Conclusions

In patients with severe and persistent bleeding, changing the ECMO circuit decreased clinical bleeding and transfusion needs and increased platelet and fibrinogen levels. Oxygenation parameters were not significantly modified after the change in the group with thrombosis.

Availability of data and materials

All data can be accessed via the corresponding author.

References

  1. Blum JM, Lynch WR, Coopersmith CM. Clinical and billing review of extracorporeal membrane oxygenation. Chest. 2015;147(6):1697–703.

    Article  PubMed  Google Scholar 

  2. Thiagarajan RR, Barbaro RP, Rycus PT, Mcmullan DM, Conrad SA, Fortenberry JD, et al. Extracorporeal life support organization registry international report 2016. ASAIO J. 2017;63(1):60–7.

    Article  PubMed  Google Scholar 

  3. McMichael ABV, Ryerson LM, Ratano D, Fan E, Faraoni D, Annich GM. 2021 ELSO adult and pediatric anticoagulation guidelines. ASAIO J. 2022;68(3):303–10.

    Article  PubMed  Google Scholar 

  4. Murphy DA, Hockings LE, Andrews RK, Aubron C, Gardiner EE, Pellegrino VA, et al. Extracorporeal membrane oxygenation-hemostatic complications. Transfus Med Rev. 2015;29(2):90–101.

    Article  PubMed  Google Scholar 

  5. Lubnow M, Philipp A, Foltan M, Bull Enger T, Lunz D, Bein T, et al. Technical complications during veno-venous extracorporeal membrane oxygenation and their relevance predicting a system-exchange–retrospective analysis of 265 cases. PLoS ONE. 2014;9(12): e112316.

    Article  PubMed  PubMed Central  Google Scholar 

  6. Kortchinsky T, Mussot S, Rezaiguia S, Artiguenave M, Fadel E, Stephan F. Extracorporeal life support in lung and heart-lung transplantation for pulmonary hypertension in adults. Clin Transplant. 2016;30(9):1152–8.

    Article  CAS  PubMed  Google Scholar 

  7. Lansink-Hartgring AO, Hoffmann R, van den Bergh W, de Vries A. Changes in red blood cell properties and platelet function during extracorporeal membrane oxygenation. J Clin Med. 2020;9(4):E1168.

    Article  Google Scholar 

  8. Appelt H, Philipp A, Mueller T, Foltan M, Lubnow M, Lunz D, et al. Factors associated with hemolysis during extracorporeal membrane oxygenation (ECMO)-Comparison of VA- versus VV ECMO. PLoS ONE. 2020;15(1): e0227793.

    Article  CAS  PubMed  PubMed Central  Google Scholar 

  9. Cartwright B, Bruce HM, Kershaw G, Cai N, Othman J, Gattas D, et al. Hemostasis, coagulation and thrombin in venoarterial and venovenous extracorporeal membrane oxygenation: the HECTIC study. Sci Rep. 2021;11(1):7975.

    Article  CAS  PubMed  PubMed Central  Google Scholar 

  10. Millar JE, Fanning JP, McDonald CI, McAuley DF, Fraser JF. The inflammatory response to extracorporeal membrane oxygenation (ECMO): a review of the pathophysiology. Crit Care. 2016;20(1):387.

    Article  PubMed  PubMed Central  Google Scholar 

Download references

Acknowledgements

Not applicable.

Funding

Support was provided solely by institutional sources. Drs Genty, Burguburu, Imbert, Roman, Wirth, Thès, and Stéphan have no conflicts of interest to disclose.

Author information

Authors and Affiliations

Authors

Contributions

TG, JT, and FS were involved in conceptualization. TG and SB were involved in investigation. TG and FS were involved in writing—original draft and writing—review and editing and took responsibility for all aspects of the work and for ensuring that questions related to the accuracy or integrity of any part of the work are appropriately investigated and resolved. All authors read and approved the final version of the manuscript submitted for publication.

Corresponding author

Correspondence to Thibaut Genty.

Ethics declarations

Ethical approval and consent to participate

Procedures were followed in accordance with the ethical standards of the responsible committee on human experimentation and with the Helsinki Declaration of 1975. A national review board approved the study (CERAR IRB00010254 accepted on 09.08.2021 with “Evolution des paramètres biologiques et cliniques au décours des changements de circuits d’ECMO” as study title. All patients received an information and no-objection statements in accordance with the French law on retrospective studies of de-identified health data.

Consent for publication

Not applicable.

Competing interests

The authors declare that they have no competing interests.

Additional information

Publisher's Note

Springer Nature remains neutral with regard to jurisdictional claims in published maps and institutional affiliations.

Rights and permissions

Open Access This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.

Reprints and permissions

About this article

Check for updates. Verify currency and authenticity via CrossMark

Cite this article

Genty, T., Burguburu, S., Imbert, A. et al. Circuit change during extracorporeal membrane oxygenation: single-center retrospective study of 48 changes. Crit Care 27, 219 (2023). https://doi.org/10.1186/s13054-023-04503-9

Download citation

  • Received:

  • Accepted:

  • Published:

  • DOI: https://doi.org/10.1186/s13054-023-04503-9

Keywords