- Research
- Open access
- Published:
Effects of an asymmetrical high flow nasal cannula interface in hypoxemic patients
Critical Care volume 27, Article number: 145 (2023)
Abstract
Background
Optimal noninvasive respiratory support for patients with hypoxemic respiratory failure should minimize work of breathing without increasing the transpulmonary pressure. Recently, an asymmetrical high flow nasal cannula (HFNC) interface (Duet, Fisher & Paykel Healthcare Ltd), in which the caliber of each nasal prong is different, was approved for clinical use. This system might reduce work of breathing by lowering minute ventilation and improving respiratory mechanics.
Methods
We enrolled 10 patients ≥ 18 years of age who were admitted to the Ospedale Maggiore Policlinico ICU in Milan, Italy, and had a PaO2/FiO2 < 300 mmHg during HFNC support with a conventional cannula. We investigated whether the asymmetrical interface, compared to a conventional high flow nasal cannula, reduces minute ventilation and work of breathing. Each patient underwent support with the asymmetrical interface and the conventional interface, applied in a randomized sequence. Each interface was provided at a flow rate of 40 l/min followed by 60 l/min. Patients were continuously monitored with esophageal manometry and electrical impedance tomography.
Results
Application of the asymmetrical interface resulted in a −13.5 [−19.4 to (−4.5)] % change in minute ventilation at a flow rate of 40 l/min, p = 0.006 and a −19.6 [−28.0 to (−7.5)] % change at 60 l/min, p = 0.002, that occurred despite no change in PaCO2 (35 [33–42] versus 35 [33–43] mmHg at 40 l/min and 35 [32–41] versus 36 [32–43] mmHg at 60 l/min). Correspondingly, the asymmetrical interface lowered the inspiratory esophageal pressure–time product from 163 [118–210] to 140 [84–159] (cmH2O*s)/min at a flow rate of 40 l/min, p = 0.02 and from 142 [123–178] to 117 [90–137] (cmH2O*s)/min at a flow rate of 60 l/min, p = 0.04.
The asymmetrical cannula did not have any impact on oxygenation, the dorsal fraction of ventilation, dynamic lung compliance, or end-expiratory lung impedance, suggesting no major effect on PEEP, lung mechanics, or alveolar recruitment.
Conclusions
An asymmetrical HFNC interface reduces minute ventilation and work of breathing in patients with mild-to-moderate hypoxemic respiratory failure supported with a conventional interface. This appears to be primarily driven by increased ventilatory efficiency due to enhanced CO2 clearance from the upper airway.
Introduction
High flow nasal cannula (HFNC) is the recommended first-line noninvasive respiratory support for patients with acute hypoxemic respiratory failure [1,2,3]. Despite its simple design, HFNC is associated with clinically relevant physiological benefits, promotes lung and diaphragm protection, and potentially allows clinicians to avoid endotracheal intubation [4]. Several mechanisms have been extensively described: greater concordance between set FiO2 and alveolar PO2 improves oxygenation; positive end-expiratory pressure (PEEP) effect increases end-expiratory lung volume [5]; and CO2 washout from the upper airway decreases minute ventilation [5, 6]. All may increase patient comfort compared to standard, low flow oxygen therapy. Although the presence and magnitude of these physiological effects vary from patient to patient, their combination decreases excessive respiratory drive. The final clinically relevant mechanisms being 1) decreased inspiratory effort and risk of diaphragmatic injury [7] and 2) decreased transpulmonary driving pressure and risk of patient self-inflicted lung injury [8].
However, the percentage of patients failing HFNC support is still high [9, 10]. Failure of HFNC leads to escalation of noninvasive support (to helmet continuous positive airway pressure and/or noninvasive positive pressure ventilation) and admission to the ICU. Delayed intubation after HFNC failure is associated with increased mortality [11]. Thus, simple methods to increase the efficacy of HFNC support would be a welcome addition to our treatments to avoid failure and/or admission to the ICU. Recently, we described additional physiological benefits conferred by HFNC delivered at flows higher than 60Â l/min, but patient comfort was poor, and sedation might be required to facilitate this approach [12]. Alternatively, awake prone positioning during support with HFNC could further increase oxygenation [13,14,15], maximize lung protection [16], and decrease work of breathing [17]. However, prone positioning requires active patient collaboration and might not be feasible in postoperative respiratory failure and severely obese patients.
Recently, a novel HFNC interface that uses an asymmetrical cannula design was approved for clinical use. The asymmetrical cannula features one prong with a smaller diameter and one prong with a larger diameter, resulting in an overall 30–40% increase in the total cross-sectional area compared to an equally sized conventional interface [18]. Only bench studies have been conducted so far, demonstrating that the asymmetrical configuration potentially increases the positive airway pressure effect and enhances CO2 washout compared to a conventional symmetrical cannula [18, 19]. Airway pressure may be increased with the asymmetrical interface due to the overall greater total cross-sectional area of the prongs, resulting in a greater prong area-to-nare area ratio [6] and CO2 washout may be augmented due to the creation of a pressure difference between the nasal cavities and a pattern of reverse flow from the more occluded to the less occluded nare [18].
In the present study, we investigated the impact of the asymmetrical cannula on minute ventilation, work of breathing and the underlying physiological mechanisms in spontaneously breathing patients with acute mild-to-moderate hypoxemic respiratory failure.
Methods
This was a single-center, prospective, physiologic, crossover study of patients admitted to the intensive care unit (ICU) of the Ospedale Maggiore Policlinico in Milan, Italy. Patients ≥ 18 years of age were eligible for inclusion if they were admitted to the ICU due to an acute respiratory condition starting within the prior 7 days and a PaO2/FiO2 < 300 mmHg while already supported with a conventional HFNC interface. Exclusion criteria included pregnancy, respiratory acidosis of any origin (PaCO2 > 45 mmHg and pH < 7.30), presence of neuromuscular disease, and contraindications to esophageal pressure monitoring (uncontrolled coagulopathy, nasal trauma, esophageal disease). The study protocol was approved by the research ethics board of the Ospedale Maggiore Policlinico (Ref. no. 930_2022).
After obtaining informed consent, baseline demographic and clinical data were collected, and an esophageal balloon catheter (Cooper Surgical, Trumbull, USA) was inserted and calibrated according to the presence of cardiac oscillations and negative deflections during inspiration [5, 20]. Then, a 16-electrode electrical impedance tomography (EIT) belt was placed around the chest at the fifth or sixth intercostal space and connected to a bedside EIT monitor. Esophageal pressure (Pes) and EIT data were recorded simultaneously at 20 Hz and stored on a dedicated data acquisition system (Dräger PulmoVista® 500, Lübeck, Germany) for offline analysis.
The study protocol consisted of application of the asymmetrical interface (Optiflow Duet, Fisher and Paykel Healthcare, Auckland, New Zealand) and the conventional interface, in a randomized sequence, for 30Â min each. The size of the conventional interface was the one in place for clinical use at study enrollment. The asymmetrical interface was sized identically to the conventional interface in place at enrollment (i.e., large or medium size for both). No instructions were provided to patients regarding mouth opening during breathing.
Each interface was provided at a flow rate of 40 l/min for 15 min followed by a flow rate of 60 l/min for 15 min (4 total steps with randomization of cannula type). An oxygen saturation (SpO2) of 92–96% was maintained throughout the study by adjusting the set FiO2. In the last 3–5 min of each step, we recorded hemodynamic and respiratory parameters, Pes and EIT data, assessed dyspnea, and obtained arterial blood gas analyses in patients with an arterial cannula. Tidal volume (VT) was estimated from the tidal change in impedance using a conversion factor as in our previous work [21].
The primary endpoint was the change in minute ventilation in patients supported with the asymmetrical cannula interface compared to the standard interface at each flow rate. Secondary endpoints included the impact on work of breathing, quantified as the simplified inspiratory pressure time product per minute (PTPmin), calculated from the esophageal pressure waveform, and the following physiological parameters, chosen to identify potential mechanisms explaining the effects of the asymmetrical interface:
Gas exchange
-
1.
PaCO2
-
2.
SpO2/FiO2
-
3.
PaO2/FiO2
Lung mechanics
-
1.
Dynamic lung compliance computed as VT / ∆Pes [17]
-
2.
Regional dynamic lung compliance (ventral and dorsal)
-
3.
End-expiratory lung volume analyzed as the end-expiratory lung impedance
-
4.
The percentage of ventilation reaching the dependent half of the lungs (dorsal fraction of ventilation)
Indicators of respiratory drive
-
1.
Mean inspiratory flow computed as VT / inspiratory time
-
2.
Modified Borg dyspnea scale [22]
-
3.
Ratio of oxygen saturation (ROX) index [10]
-
4.
Respiratory rate
We also investigated whether the asymmetrical cannula had any effect on hemodynamics by evaluating the following:
Hemodynamics
-
1.
Systolic blood pressure
-
2.
Mean arterial blood pressure
-
3.
Heart rate
Statistical Analysis
Based on prior work [5] and a predicted 20% change in minute ventilation considered as clinically relevant, a sample size of 8 patients was selected to attain a probability of type-I error of 0.05 and a statistical power of 80%. The sample size was increased to 10 patients to account for potential missing data (i.e., poor quality of EIT recordings). EIT data were obtained and analyzed in all patients, Pes data were obtained from 8 patients, and arterial blood gas analyses were obtained in 5 patients.
Descriptive statistics were used to characterize the study population. Continuous data are presented as median and interquartile range [IQR]. Comparisons between interfaces were analyzed at both flow rates using the Wilcoxon signed rank test. A p-value of < 0.05 was considered statistically significant.
To investigate whether baseline clinical or physiological characteristics predicted the magnitude of reduction in minute ventilation after application of the asymmetrical cannula, we performed simple linear regression analysis of relevant predictors and the percentage change in minute ventilation resulting from application of the asymmetrical cannula at each flow rate. We considered the following predictors: PaO2/FiO2, number of quadrants affected on chest radiograph, and Sequential Organ Failure Assessment (SOFA) score at enrollment, Simplified Acute Physiology Score (SAPS II) score at ICU admission, respiratory rate, baseline minute ventilation and mean inspiratory flow (respiratory drive) measured during HFNC support with the conventional interface at 40Â l/min. When a significant association was identified for a given predictor, we performed a linear mixed-effects analysis that accounted for the repeated-measures design across flow rates. The percentage change in minute ventilation was defined as the dependent variable, flow rate and the predictor were considered fixed-effects variables, and individual patients were treated as a random-effects variable.
Results
Data from 10 patients were analyzed and their baseline demographics, and clinical information is presented in Table 1. Patients were majoritarily male and had a SAPS II score of 41 [37–47] upon ICU admission. At enrollment, the PaO2/FiO2 was 249 [187–258] mmHg and 2 [1–4] quadrants were affected on chest radiography.
Application of the asymmetrical interface resulted in a decrease in the minute ventilation from 10.9 [7.4–13.7] to 9.7 [6.3–10.8] l/min at a flow rate of 40 l/min, p = 0.01 and from 9.4 [6.5–14.7] to 8.5 [4.8–10.8] l/min at a flow rate of 60 l/min, p = 0.002. This corresponded to a -13.5 [−19.4 to (−4.5)] % change in minute ventilation at a flow rate of 40 l/min, p = 0.006 (Fig. 1) and a −19.6 [−28.0 to (−7.5)] % change at 60 l/min, p = 0.002 (Fig. 2). The change in minute ventilation was mediated by a decrease in tidal volume (512 [356–647] to 401 [336–600] ml at a flow rate of 40 l/min, p = 0.004 and 468 [320–662] to 396 [244–569] ml at 60 l/min, p = 0.02) that occurred despite no change in PaCO2 (35 [33–42] versus 35 [33–43] mmHg at 40 l/min and 35 [32–41] versus 36 [32–43] mmHg at 60 l/min) (Figs. 1 and 2, Table 2).
Correspondingly, application of the asymmetrical interface lowered the inspiratory esophageal pressure–time product (an index of the metabolic work of breathing over 1 min) from 163 [118–210] to 140 [84–159] (cmH2O*s)/min at a flow rate of 40 l/min, p = 0.02 and from 142 [123–178] to 117 [90–137] (cmH2O*s)/min at a flow rate of 60 l/min, p = 0.04 (Figs. 1 and 2). Interestingly, 2 patients who manifested no change or an increase in minute ventilation seen in Fig. 1 also developed a decrease in esophageal pressure–time product, suggesting that minute ventilation might not always be an accurate reflection of work of breathing. There was a trend toward a decrease in respiratory rate (Additional file 1: Fig. S1) and the change in Pes (∆Pes) (Additional file 1: Fig. S2) that did not reach statistical significance at each flow rate.
Application of the asymmetrical cannula did not have any impact on average values of oxygenation, the dorsal fraction of ventilation, global or regional dynamic lung compliance, or end-expiratory lung impedance (Table 2), suggesting that there was no major effect on PEEP, lung mechanics, or alveolar recruitment at a flow rate of 40 l/min or 60 l/min (Fig. 3). Notably, 4 patients (40%) did manifest an increase in end-expiratory lung volume with the asymmetrical interface at 40 l/min (Fig. 3) and 6 patients (60%) at 60 l/min (Fig. 3), suggesting that a PEEP effect may be present in specific patient subgroups (e.g., extrapulmonary lung injury). There was no effect on patient reported dyspnea, and the asymmetrical interface was well tolerated overall.
Of the tested clinical and physiological predictors, higher SAPS II score at ICU admission was significantly associated with a greater percentage decrease in minute ventilation conferred by the asymmetrical interface at both flow rates (Fig. 4), whereas higher minute ventilation (R2 = 0.46, p = 0.03) and mean inspiratory flow during support with the conventional interface were associated with a greater decrease in minute ventilation at a flow rate of 40 l/min (Fig. 4). Linear mixed-effects analysis demonstrated that every 1-point increase in SAPS II at admission was associated with a −1.1 [−1.9 to (−0.38)] percentage point change in minute ventilation (Table 3) suggesting that greater severity of illness at ICU admission was associated with a greater reduction in minute ventilation following application of the asymmetrical cannula. Linear mixed-effects analyses of baseline minute ventilation and mean inspiratory flow did not demonstrate a significant relationship with the change in minute ventilation.
Discussion
The main finding of our study is that application of an asymmetrical high flow nasal cannula interface in patients with mild-to-moderate hypoxemia who are already supported with a conventional interface is associated with a reduction in minute ventilation and work of breathing despite no change in PaCO2, indicating increased ventilatory efficiency. End-expiratory lung inflation and alveolar recruitment, on average, were not improved by the asymmetrical interface suggesting absence of large PEEP effect compared to the conventional interface.
These results are consistent with findings recently reported by Tatkov et al. who performed a study of the asymmetrical interface in an upper airway model [18]. They demonstrated improved dead-space clearance and more efficient CO2 elimination from the upper airway with the asymmetrical interface that was due to the preferential flow of gas out of the lesser occluded nare and redirected flow from the larger cannula to the contralateral nare during expiration. Similar findings were also recently reported by Vieira et al. in a bench study comparing a conventional interface to the asymmetrical interface [19]. They demonstrated a reduction in rebreathing volume from the upper airway with the asymmetrical interface that was generally amplified at higher flow rates. The potential benefits for patients being decreased dyspnea and work of breathing.
Our data demonstrate that the asymmetrical interface was not associated with any change in oxygenation or lung mechanics, suggesting that the overall mechanism of the reduction in work of breathing may not be due to a mean airway pressure effect on lung recruitment. This finding differs from data presented in previous studies. Studying symmetrical interfaces of varying size, Pinkham et al. demonstrated that a greater prong area-to-nare area ratio resulted in greater end-expiratory airway pressure in a bench model and that this effect was magnified at higher flow rates [23]. Tatkov et al. reported higher PEEP during support with the asymmetrical interface, particularly at flow rates of 40 and 60 l/min with a maximal difference of 2–3 cmH2O [18]. Vieira et al. also reported higher nasopharyngeal pressure at end expiration with the asymmetrical cannula. The difference in pressure was more pronounced with increasing flow rate and respiratory rate. The authors of both studies attributed this difference to increased airflow resistance induced by the asymmetrical interface, both during inspiration and expiration [19, 24]. We did not measure end-expiratory airway pressure directly and cannot corroborate these findings. A lack of evidence supporting an airway pressure effect in our study may be explained by a lack of recruitability in this patient population or by an increase in airway pressure too small to be clinically relevant. Another possibility is that we did not instruct patients not to breathe with the mouth open, a common occurrence in awake patients. We, therefore, cannot make a conclusion regarding the effect of the asymmetrical interface on PEEP. However, if in some patients, PEEP was increased, the potential physiological benefits of the asymmetrical cannula could be even greater.
The finding that a greater severity of illness at ICU admission, assessed with the SAPS II score, might predict a greater reduction in minute ventilation following application of the asymmetrical cannula suggests that patients with a greater burden of systemic disease may derive physiologic benefits from support with the asymmetrical interface. This complements our prior study that demonstrated that support with HFNC was effective at reducing respiratory drive and work of breathing in patients with extrapulmonary sepsis, over 50% of whom had a diagnosis of septic shock with a median SAPS II score of 37 at admission to the ICU [25]. In patients with more severe systemic disease and increased CO2 production, enhanced CO2 clearance by the asymmetrical cannula could be even more clinically relevant. However, our finding should also be considered in the context of a recent retrospective study of 200 patients with COVID-19, in whom a greater SAPS II score was associated with HFNC failure, defined as a need for intubation following admission to the ICU [11] and a study of 202 onco-hematology patients who received HFNC support in the ICU for acute respiratory failure, in whom a greater SAPS II score was associated with increased risk of intubation and hospital mortality after HFNC failure [26]. These studies underscore the clinical need for a HFNC device with increased physiological effects in patients with a greater severity of illness.
The exploratory finding that higher minute ventilation and mean inspiratory flow (VT / inspiratory time), reflecting higher respiratory drive, during support with the conventional interface may indicate a patient who will manifest a greater reduction in minute ventilation and work of breathing with the asymmetrical interface may help clinicians in tailoring their approach to selecting patients most likely to benefit from the asymmetrical interface. Previous data have shown that higher PaCO2 during support with low flow oxygen therapy predicted a greater reduction in the tidal change in Pes following support with HFNC [5].
Overall, we observed that the impact of the asymmetrical interface was greatest when it was applied at 60 l/min. Compared to the conventional interface applied at 40 l/min, the asymmetrical interface applied at 60 l/min resulted in a minute ventilation that was > 2 l/min lower. Despite our patients having a normal PaCO2 and pH during support with the conventional interface, enhanced CO2 clearance by the asymmetrical interface may have assisted patients in maintaining a normal PaCO2 with unchanged respiratory drive at a HFNC flow rate of 40 l/min and decreased respiratory drive at 60 l/min [27, 28]. Reducing elevated respiratory drive and breathing effort while maintaining lung recruitment and gas exchange represent key objectives of a lung and diaphragm-protective ventilation strategy [7].
Our study has limitations: First a small number of patients were studied and only 5 had measurements of PaCO2 available at each step, rendering conclusions regarding ventilatory efficiency hypothesis generating; second, all patients had mild-to-moderate hypoxemia without significant dyspnea, limiting the generalizability of the findings to a broader critically ill and hypercapnic population; third, we did not control for the potential impact of open mouth breathing, a variable that has a significant impact on the development of PEEP during support with HFNC [24]; fourth, we did not measure gastric pressures or assess for expiratory muscle recruitment that could have led to an overestimation of ∆Pes and the esophageal pressure–time product; and fifth, we estimated VT and minute ventilation based on EIT derived measurements of the tidal change in lung impedance. However, this method has been used in previous studies [25] and has increased validity given that patients are compared with themselves across HFNC interface steps.
Conclusion
An asymmetrical HFNC interface reduces minute ventilation and work of breathing in patients with mild-to-moderate hypoxemic respiratory failure supported with a conventional high flow nasal cannula. This appears to be primarily driven by increased ventilatory efficiency due to enhanced CO2 clearance from the upper airway.
Availability of data and materials
The datasets used and analyzed in the current study are available from the corresponding author on reasonable request.
Abbreviations
- HFNC:
-
High flow nasal cannula
- PEEP:
-
Positive end-expiratory pressure
- ICU:
-
Intensive care unit
- EIT:
-
Electrical impedance tomography
- Pes:
-
Esophageal pressure
- ∆Pes:
-
Change in esophageal pressure
- VT :
-
Tidal volume
- PTPmin :
-
Inspiratory esophageal pressure time product per minute
- ROX index:
-
Ratio of oxygen saturation index
- SOFA:
-
Sequential organ failure assessment
- SAPS II:
-
Simplified acute physiology score
- ∆Z:
-
Impedance change
References
Rochwerg B, Einav S, Chaudhuri D, Mancebo J, Mauri T, Helviz Y, et al. The role for high flow nasal cannula as a respiratory support strategy in adults: a clinical practice guideline. Intensive Care Med. 2020;46(12):2226–37.
Grieco DL, Maggiore SM, Roca O, Spinelli E, Patel BK, Thille AW, et al. Non-invasive ventilatory support and high-flow nasal oxygen as first-line treatment of acute hypoxemic respiratory failure and ARDS. Intensive Care Med. 2021;47(8):851–66.
Mauri T, Wang YM, Dalla Corte F, Corcione N, Spinelli E, Pesenti A. Nasal high flow: physiology, efficacy and safety in the acute care setting, a narrative review. Open Access Emerg Med. 2019;11:109–20.
Rochwerg B, Granton D, Wang DX, Helviz Y, Einav S, Frat JP, et al. High flow nasal cannula compared with conventional oxygen therapy for acute hypoxemic respiratory failure: a systematic review and meta-analysis. Intensive Care Med. 2019;45(5):563–72.
Mauri T, Turrini C, Eronia N, Grasselli G, Volta CA, Bellani G, et al. Physiologic effects of high-flow nasal cannula in acute hypoxemic respiratory failure. Am J Respir Crit Care Med. 2017;195(9):1207–15.
Pinkham M, Tatkov S. Effect of flow and cannula size on generated pressure during nasal high flow. Crit Care. 2020;24(1):248.
Goligher EC, Dres M, Patel BK, Sahetya SK, Beitler JR, Telias I, et al. Lung- and diaphragm-protective ventilation. Am J Respir Crit Care Med. 2020;202(7):950–61.
Brochard L, Slutsky A, Pesenti A. Mechanical ventilation to minimize progression of lung injury in acute respiratory failure. Am J Respir Crit Care Med. 2017;195(4):438–42.
Roca O, Messika J, Caralt B, Garcia-de-Acilu M, Sztrymf B, Ricard JD, et al. Predicting success of high-flow nasal cannula in pneumonia patients with hypoxemic respiratory failure: the utility of the ROX index. J Crit Care. 2016;35:200–5.
Roca O, Caralt B, Messika J, Samper M, Sztrymf B, Hernandez G, et al. An index combining respiratory rate and oxygenation to predict outcome of nasal high-flow therapy. Am J Respir Crit Care Med. 2019;199(11):1368–76.
Ait Hamou Z, Levy N, Charpentier J, Mira JP, Jamme M, Jozwiak M. Use of high-flow nasal cannula oxygen and risk factors for high-flow nasal cannula oxygen failure in critically-ill patients with COVID-19. Respir Res. 2022;23(1):329.
Basile MC, Mauri T, Spinelli E, Dalla Corte F, Montanari G, Marongiu I, et al. Nasal high flow higher than 60 L/min in patients with acute hypoxemic respiratory failure: a physiological study. Crit Care. 2020;24(1):654.
Tu GW, Liao YX, Li QY, Dong H, Yang LY, Zhang XY, et al. Prone positioning in high-flow nasal cannula for COVID-19 patients with severe hypoxemia: a pilot study. Ann Transl Med. 2020;8(9):598.
Ding L, Wang L, Ma W, He H. Efficacy and safety of early prone positioning combined with HFNC or NIV in moderate to severe ARDS: a multi-center prospective cohort study. Crit Care. 2020;24(1):28.
Despres C, Brunin Y, Berthier F, Pili-Floury S, Besch G. Prone positioning combined with high-flow nasal or conventional oxygen therapy in severe Covid-19 patients. Crit Care. 2020;24(1):256.
Riera J, Perez P, Cortes J, Roca O, Masclans JR, Rello J. Effect of high-flow nasal cannula and body position on end-expiratory lung volume: a cohort study using electrical impedance tomography. Respir Care. 2013;58(4):589–96.
Chiumello D, Chiodaroli E, Coppola S, Cappio Borlino S, Granata C, Pitimada M, et al. Awake prone position reduces work of breathing in patients with COVID-19 ARDS supported by CPAP. Ann Intensive Care. 2021;11(1):179.
Tatkov S, Rees M, Gulley A, van den Heuij LG, Nilius G. Asymmetrical nasal high flow ventilation improves clearance of CO2 from the anatomical dead space and increases positive airway pressure. J Appl Physiol. 2023;134(2):365–77.
Vieira F, Rodrigues A, Schreiber A, Brault C, Subira C, Brochard L, editors. Comparing physiological effects of standard vs asymmetrical high-flow nasal cannula: a bench study. Canadian Critical Care Forum; 2022 2022; Toronto, Canada.
Akoumianaki E, Maggiore SM, Valenza F, Bellani G, Jubran A, Loring SH, et al. The application of esophageal pressure measurement in patients with respiratory failure. Am J Respir Crit Care Med. 2014;189(5):520–31.
Mauri T, Alban L, Turrini C, Cambiaghi B, Carlesso E, Taccone P, et al. Optimum support by high-flow nasal cannula in acute hypoxemic respiratory failure: effects of increasing flow rates. Intensive Care Med. 2017;43(10):1453–63.
Johnson MJ, Close L, Gillon SC, Molassiotis A, Lee PH, Farquhar MC, et al. Use of the modified Borg scale and numerical rating scale to measure chronic breathlessness: a pooled data analysis. Eur Respir J. 2016;47(6):1861–4.
Pinkham MI, Domanski U, Franke KJ, Hartmann J, Schroeder M, Williams T, et al. Effect of respiratory rate and size of cannula on pressure and dead-space clearance during nasal high flow in patients with COPD and acute respiratory failure. J Appl Physiol. 2022;132(2):553–63.
Vieira F, Bezerra FS, Coudroy R, Schreiber A, Telias I, Dubo S, et al. High-flow nasal cannula compared with continuous positive airway pressure: a bench and physiological study. J Appl Physiol. 2022;132:1580–90.
Mauri T, Spinelli E, Pavlovsky B, Grieco DL, Ottaviani I, Basile MC, et al. Respiratory drive in patients with sepsis and septic shock: modulation by high-flow nasal cannula. Anesthesiology. 2021;135(6):1066–75.
Saillard C, Lambert J, Tramier M, Chow-Chine L, Bisbal M, Servan L, et al. High-flow nasal cannula failure in critically ill cancer patients with acute respiratory failure: moving from avoiding intubation to avoiding delayed intubation. PLoS ONE. 2022;17(6):e0270138.
Vaporidi K, Akoumianaki E, Telias I, Goligher EC, Brochard L, Georgopoulos D. Respiratory drive in critically ill patients. Pathophysiology and clinical implications. Am J Respir Crit Care Med. 2020;201(1):20–32.
Spinelli E, Mauri T, Beitler JR, Pesenti A, Brodie D. Respiratory drive in the acute respiratory distress syndrome: pathophysiology, monitoring, and therapeutic interventions. Intensive Care Med. 2020;46(4):606–18.
Acknowledgements
Not applicable
Funding
This study was supported by research funding from Fisher and Paykel.
Author information
Authors and Affiliations
Contributions
TM conceived of the work. DS, TM, and ES collected and interpreted patient data. DS and TM wrote the manuscript. DS, TM, ES, SC, AL, AG, and GG revised the manuscript. All authors read and approved the final manuscript.
Corresponding author
Ethics declarations
Ethics approval and consent to participate
The study protocol was approved by the research ethics board of the Ospedale Maggiore Policlinico (Ref. no. 930_2022). Consent for publication of data was obtained from all patients.
Competing interests
DS, ES, SC, AL, and AG report no conflicts of interest. GG reports receiving personal fees from Getinge, Biotest, Pfizer, Dräger, and Fisher and Paykel. TM reports receiving personal fees from Fisher and Paykel, Dräger, Mindray, and B. Braun.
Additional information
Publisher's Note
Springer Nature remains neutral with regard to jurisdictional claims in published maps and institutional affiliations.
Supplementary Information
Additional file 1
: Fig. S1. Respiratory rate before and after application of the asymmetric high flow nasal cannula interface. The difference was not significant at each flow rate (see Table 2 in main manuscript for statistical information). Horizontal bars represent median and interquartile range. Fig. S2. The change in esophageal pressure (∆Pes) before and after application of the asymmetric high flow nasal cannula interface. The difference was not significant at each flow rate (see Table 2 in main manuscript for statistical information). Horizontal bars represent median and interquartile range.
Rights and permissions
Open Access This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.
About this article
Cite this article
Slobod, D., Spinelli, E., Crotti, S. et al. Effects of an asymmetrical high flow nasal cannula interface in hypoxemic patients. Crit Care 27, 145 (2023). https://doi.org/10.1186/s13054-023-04441-6
Received:
Accepted:
Published:
DOI: https://doi.org/10.1186/s13054-023-04441-6