Although uncommon, oxygenator thrombosis in extracorporeal membrane oxygenation (ECMO) devices remains a potential reason for exchanging the circuit, the risk of thrombosis increases with the duration of ECMO therapy and with procoagulant conditions, such as COVID-19 [1, 2]. With a worldwide increased demand for ECMO support, alternative methods to prolong lifespan of the oxygenator have been considered. Recombinant tissue-type plasminogen activator (rtPA) could be a potential approach in this setting [3].
From March 1, 2020, to March 30, 2022, 3 patients from a population of 99 were treated with either veno-arterial (V-A) and veno-venous (V-V) ECMO or extra-corporeal CO2 removal (ECCO2R) support in our Intensive Care Unit (ICU) required the injection of rtPA for acute oxygenator thrombosis. Characteristics of the study cohort are reported in Table 1. Informed consent was waived by our Ethics Committee because the data collection was retrospective from the data patient monitoring system (PDMS). The decision to administer rtPA was discussed and decided in the multidisciplinary medical round.
The first patient was on V-V ECMO during the first COVID-19 wave and was fully dependent on the ECMO support because of severe respiratory failure; after 4 days on ECMO, the oxygenator showed rapid signs of clotting (i.e., ECMO flow reduced to 3 L/min) and arterial oxygenation dropped to 55%. Because of no chance of survival considering the time required to have a perfusionist available, the administration of low-dose rtPA (5 mg bolus followed by additional 5 mg after 5 min) into the drainage section of the oxygenator was decided; within 15 min, the ECMO flow returned to the initial baseline values. Anticoagulation was adjusted to higher targets and no recurrence of oxygenator clotting was observed thereafter. The second patient presented severe respiratory acidemia (i.e., pH 7.18; PaCO2 83 mmHg) with protective lung ventilation requiring ECCO2R; within 3 h from onset of therapy, oxygenator thrombosis occurred in non-working hours and it was decided to administer rtPA to avoid complete circuit and cannula clotting. Drug administration helped to dissolve the clot and result in a normal function of the device. Similar results were observed in the third patient; no hemorrhagic complication was observed within the 5 days following rtPA injection. Heparin was not interrupted during thrombolysis.
This is the second report on the use of “off-label” rtPA for oxygenator thrombosis in ECMO patients. In another small cohort, 4 patients were successfully treated with several administrations of rtPA [3], without significant complications; however, all patients had only moderately impaired systemic oxygenation, not justifying immediate oxygenator exchange, while in our cohort two patients had life-threatening respiratory compromise with severe hypoxemia. Unfractionated heparin (UFH) remains the anticoagulant of choice for patients on ECMO, although the optimal anticoagulation target and the most effective coagulation parameter to use to adjust heparin regimens remain unknown in these patients. The reason we decided to continue with UFH was due to the relatively low anti-Xa values at clotting, which suggest a potentially insufficient anticoagulation regimen rather than the ineffectiveness of UFH. Moreover, the indication for oxygenator exchange in case of suspected thrombosis is also undetermined and might be based on membrane performance, hemolysis and/or patient’s status [4]. Further studies are required to understand the timing, dose and potential complications of rtPA use for acute oxygenator thrombosis in ECMO patients. Our findings suggest that rtPA given at lower doses than thrombolysis for pulmonary embolism or ischemic stroke (i.e., 5–20 mg) could be effective to treat life-threatening oxygenator thrombosis.
Availability of data and materials
On request to the corresponding author.
References
Figueroa Villalba CA, McMullan DM, Reed RC, Chandler WL. Thrombosis in extracorporeal membrane oxygenation (ECMO) circuits. ASAIO J. 2022;68(8):1083–92.
Barbaro RP, MacLaren G, Boonstra PS, et al. Extracorporeal Life Support Organization: extracorporeal membrane oxygenation support in COVID-19: an international cohort study of the Extracorporeal Life Support Organization registry. Lancet. 2020;396:1071–8.
Mang S, Danziger G, Metz C, Rixecker T, Becker A, Omlor AJ, Jentgen C, Schmoll C, Seiler F, Reyher C, Muellenbach RM, Bals R, Lepper PM. Injection of recombinant tissue plasminogen activator into extracorporeal membrane oxygenators postpones oxygenator exchange in COVID-19. ASAIO J. 2022;68(8):1017–23.
Zakhary B, Vercaemst L, Mason P, Antonini MV, Lorusso R, Brodie D. How I approach membrane lung dysfunction in patients receiving ECMO. Crit Care. 2020;24(1):671.
Funding
None.
Author information
Authors and Affiliations
Contributions
FST, LN and FA wrote the main manuscript text; FST prepared the table. All authors read and approved the final manuscript.
Corresponding author
Ethics declarations
Ethical approval and consent to participate
The study has been approved by the Ethics Committee of the Erasme Hospital (P2017/263).
Competing interests
The authors declare that they have no competing interests.
Additional information
Publisher's Note
Springer Nature remains neutral with regard to jurisdictional claims in published maps and institutional affiliations.
Rights and permissions
Open Access This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.
About this article
Cite this article
Taccone, F.S., Nobile, L. & Annoni, F. Thrombolysis for ECMO oxygenator thrombosis. Crit Care 27, 142 (2023). https://doi.org/10.1186/s13054-023-04433-6
Received:
Accepted:
Published:
DOI: https://doi.org/10.1186/s13054-023-04433-6