Skip to main content

Mechanical power of ventilation and driving pressure: two undervalued parameters for pre extracorporeal membrane oxygenation ventilation and during daily management?

Abstract

The current ARDS guidelines highly recommend lung protective ventilation which include plateau pressure (Pplat < 30 cm H2O), positive end expiratory pressure (PEEP > 5 cm H2O) and tidal volume (Vt of 6 ml/kg) of predicted body weight. In contrast, the ELSO guidelines suggest the evaluation of an indication of veno-venous extracorporeal membrane oxygenation (ECMO) due to hypoxemic or hypercapnic respiratory failure or as bridge to lung transplantation. Finally, these recommendations remain a wide range of scope of interpretation. However, particularly patients with moderate-severe to severe ARDS might benefit from strict adherence to lung protective ventilation strategies. Subsequently, we discuss whether extended physiological ventilation parameter analysis might be relevant for indication of ECMO support and can be implemented during the daily routine evaluation of ARDS patients. Particularly, this viewpoint focus on driving pressure and mechanical power.

Why extended monitoring of pre-ECMO ventilation and indication might be relevant?

Several studies analyzed the potential survival benefit of extracorporeal membrane oxygenation (ECMO) in severe SARS-CoV 2 infected patients. However, the results were highly variable. International registries reported in-hospital mortality rates 90 days after initiation of ECMO therapy of 38% which is in range with the pre-SARS-CoV 2 era data [1, 2]. Recently, Whebell et al. [3] reported a very low in hospital mortality rate of 24% and an absolute reduction of mortality in the ECMO-treated patient group after propensity score matching. In contrast, others analyzed large patient cohorts encompassing 243 or 673 patients and revealed overall in hospital mortality rates up to 68% in the ECMO treated patient collective [4, 5]. Indeed, these differences might be ascribed to center experience and patient specific co-morbidities particularly age or prior immunosuppressive therapy [5]. However, the differences in outcome might also be explained by different center specific indications of ECMO support and ventilation invasiveness pre implantation. One first hint might be given by a detailed analysis of the EOLIA trial. The majority of inclusions (82%) were due to hypoxemia and the reported 60 day mortality rate was 35% in the ECMO-treated patients and 46% in the control group [6]. However, the reported mortality of the patient group who were included due to refractory acidosis by compromised protective ventilation the 60 day mortality was lower (24% in the ECMO group versus 55% in the control group) [6, 7].

The ELSO guidelines recommend veno-venous ECMO implementation after exclusion of contraindications in the following circumstances: (i) hypoxemic respiratory failure (PaO2/FiO2 < 80 mm Hg) after optimal medical management including prone positioning trial; (ii) hypercapnic respiratory failure ( pH< 7.25), despite optimal conventional mechanical ventilation (respiratory rate 35 bpm and plateau pressure ≤ 30 cm H2O) (iii) ventilatory support as a bridge to lung transplantation or primary graft dysfunction following lung transplant (Fig. 1) [8]. Conclusively, these recommendations leave plenty of room for personal interpretation.

Fig. 1
figure 1

Current guidelines for ARDS treatment [8]

In this viewpoint, we summarize the relevance of the current recommendations for acute respiratory distress syndrome (ARDS) and discuss whether extended physiological ventilation parameter analysis might improve patients` outcome. In particular, we explain the physiological and clinical relevance of driving pressure and mechanical power ventilation.

Current background of lung protective ARDS ventilation

In 1998 Dreyfuss et al. [9] reported by application of an experimental animal model, that high inflation pressure due to ventilation with high tidal volumes results in increased pulmonary edema and was reduced by straps around the chest and abdomen [10]. Apart from the avoidance of high volume ventilation, the application of positive end exspiratory pressure (PEEP) of 10 cm H20 reduced peri-vascular and alveolar edema reflecting the high relevance between over-distension and low end-expiratory lung volume for ventilator-induced lung injury (VILI) [10, 11]. These findings were approved in large human clinical trials and are the basis of the current ARDS guidelines which subsequently recommend Pplat < 30 cmH2O, PEEP > 5 cmH2O and Vt 6 ml/kg of predicted body weight (BPW) [Fig. 1] [13]. However, the Lung safe study prospectively analyzed more than 29,000 patients in 50 countries and revealed sub-optimal application of lung protective ventilation [12]. Particularly, Pplat was considered in only 40% of the ARDS patients and of these only two-thirds were ventilated in a lung protective mode (tidal volume ≤ 8 ml/kg BPW and Pplat ≤ 30 cmH2O) [12, 13]. Indeed, it is not surprising, that these single values might not be adequate for all ARDS patients. While some patients suffer severe carbon dioxide retention at Vt ventilation, others may be intolerant to high PEEP levels due to circulatory insufficiency or do not benefit from increased PEEP due to limited recruitability. Otherwise, inexactitude realization of lung protective ventilation might have serious consequences particularly for the patients with moderate-severe to severe ARDS. Lung lesions are distributed unequally and injured lung tissue or atelectasis coexists with aerated or normal lung tissue [14]. This is accompanied by a marked heterogeneity in ventilation. Particularly within the border areas between aerated and atelectatic regions up to four to five times increased stretching forces were suggested by a mathematical model application [15]. Subsequently, the potentially injurious ventilator settings were applied to a progressively smaller and more inhomogeneous “baby lung”.

What does driving pressure stand for and why might it be clinically relevant?

The applied pressure to support the delivery of the Vt is defined as driving pressure, which represents the strain applied to the lung during each ventilatory cycle. Driving pressure comprises the difference between the airway pressure at the end of the inspiration (Pplateau) and PEEP [16,17,18]. The quotient between Vt and driving pressure represents the static compliance of the respiratory system. Finally, the driving pressure reflects the Vt in relation to the compliance of the respiratory system (CRS) which is associated with ARDS severity as it reflects the proportion of lung availability for ventilation. In patients suffering ARDS, CRS was reported to be directly related to lung functional size [19,20,21].

However, the clinical benefit of moving a mere Vt to a CRS-based ventilation strategy is currently discussed. Amato et al. [22] suggested that the driving pressure was strongly associated with mortality and a decrease due to changed ventilator settings was associated with improved survival. Interestingly, this correlation was also persistent during lung protective ventilation. Recently Haudbourg et al. [19] reported that a driving pressure guided ventilation strategy with target levels between 12 and 14 cm H2O required Vt adoptions in 90% of the patients. In contrast, earlier reports suggested no significant advantage of the driving pressure concept compared to the Pplateau in view of mortality [13, 23]. However, available data is very limited and based on retrospective and observational designs or with very limited encompassed patients. Moreover, the transpulmonary driving pressure (the difference between Pplateau minus PEEP and Pesophageal-Plateau minus Pend-expiratory oesophageal) which particularly includes chest wall elastance was reported to better reflect lung stress [16, 24]. Finally, VILI was suggested to be triggered by mechanical stress and strain which is determined by Vt and endexspiratory lung volume (corresponding to higher respiratory system elastance)—both parameters are represented by the driving pressure [25]. However, driving pressure is physiologically and mathematically coupled with Vt, elastance and subsequent disease severity [26, 27]. Therefore Goligher et al. [26] analyzed the relationship between the respiratory elastance and mortality for the higher and lower Vt strategy arms. The absolute risk reduction associated with a lower Vt ventilation strategy increased progressively with increased elastance. In conclusion, driving pressure should be monitored during daily routine practice in ARDS patients and critically evaluated for Vt reductions below 6 ml/kg PBW when exceeding 15 cm H2O. Of course, the threshold is currently a matter of debate and remains to be evaluated within clinical trials. In this regard, clinicians have to be aware that PEEP changes might subsequent influence elastance (increase: overdistension; decrease: lung recruitment). Finally, clinical trials which evaluate the elastance based on very low Vt ventilation strategies potentially facilitated by extracorporeal CO2 removal strategies are urgently needed to optimize lung protection in ARDS patients.

What does  mechanical power stand for and why might it be clinically relevant?

While the relevance of the static ventilation parameter including Vt, PEEP, Pplateau and driving pressure is well established, current increasing evidence suggests a relevant contribution of the dynamic ventilation parameter including respiratory rate, inspiratory and expiratory airflow to VILI (Fig. 2) [27, 28]. Subsequently, the concept of mechanical power defined as the product of respiratory rate and total inflation energy gained attention for ventilation monitoring. Inflation energy comprises three components: (i) the power to overcome airway resistance during gas movements; (ii) the power to inflate the lung and chest wall movements, and (iii) the power to overcome end-exspiratory pressure-related recoil of the lung and respiratory system (Fig. 3) [29]. When these parameters are multiplied by the respiratory rate, mechanical power applied to the respiratory system per minute results [24]. Particularly the respiratory rate is an undervalued parameter during clinical practice. However, recent evidence revealed two ventilation strategies (High Vt; PPlateau 34 cm H2O; driving pressure 29 cm H2O versus RR 40 pb; PPlateau 17 cm H2O; driving pressure 9 cm H2O) caused the same degree of lung lesion after 48 h which suggests that also increased respiratory rate might cause increase lung injury in a specific damaged area [28, 30].

Fig. 2
figure 2

Individualized and normalized mechanical power-based and driving pressure-based ventilator management: Monitoring static ventilation parameter (Pinsp, PEEP and Vt) (purple gear) at least twice a day is highly recommended. If Pdriving exceeds 15 cm H2O, Vt reductions below 6 ml/kg PBW and PEEP reductions should be critically evaluated. For in-depth analysis MP should be determined (Fig. 3). While values below 7 J/min were reported to be save, values between 7 to 15 J/min are currently a matter of debate. MP is a function of its components (Vt, Pplateau, inspiratory Flow, RR and PEEP) and every single parameter should be evaluated for potential reductions (red gear) [40]. Normalization to the CT determined well ventilated lung tissue is highly recommended. If sufficient oxygenation and adequate CO2 removal is not realizable after optimization of the ventilator strategies, indication for ECMO and extracorporeal CO2 removal devices should be critically evaluated

Fig. 3
figure 3

Extended formula for PCV ventilation [33, 52]

Neto et al. [31] analyzed 8207 patients receiving invasive ventilation for at least 48 h hours and suggested that mechanical power in the second 25 h of ventilation might be independently associated with increased mortality of critically ill patients, a lower number of ventilator free days and survival at day 28. Concordantly, Umer et al. [32] reported that the cumulative exposure to higher intensities of mechanical ventilation was harmful and that a significant increase in the hazard of death was found to be associated with each daily increment in driving pressure and mechanical power. In contrast, Coppola et al. reported that mechanical power resulting from airway pressure and from transpulmonary pressure were assessed and not related to the outcome of ARDS patients [33]. However, both the normalization to compliance and to well-inflated tissue independently increased the intensive care mortality of 1.78 and 2.64 times for one unit increase [33].

An experimental animal setting confirmed that high mechanical power ventilation is associated with increased levels of interleukin 6 and amphiregulin expression and correlated well with diffuse alveolar damage score and club cell protein 16 extression [34]. Importantly, mechanical power combines the effects of different variables and changing of none variable may not necessarily protect the lungs, as it may increase mechanical power delivered to the lung [35]. In detail, a decreased Vt might not necessarily be translated into lung protection if respiratory rate is increased for compensation or PEEP increases may not be protective if not accompanied by declined driving pressure [30, 31, 35]. Particularly, the ARDS is very heterogenous and PEEP is regularly applied to reduce inhomogeneity. However, Maiolo et al. reported variable effects of increased airway pressure and reported an increased inhomogeneity of 20% in mild ARDS while this effect was less pronounced or even negative in severe ARDS [39]. In agreement, further evidence suggests worsened outcome due to high pressure recruitment or high PEEP levels [37].

However, the threshold of optimized mechanical power is currently a matter of debate. Based on the previous study by Neto et al. [31] 25 J/min may discriminate between higher and lower lung damage, and iso-mechanical power results in similar degree of lung damage independent of whether the reason was high Vt, respiratory rate or PEEP [28, 38]. This might be explained by an application of healthy animals but also due to exceeding the threshold causing maximal lung damage [28]. This hypothesis is in line with Cressoni et al. [39] reporting 12 J/min as threshold for mechanical power to induce VILI. Finally experimental and clinical trials to determine the optimal threshold for lung protective ventilation remain to be initiated.

Should mechanical power be normalized in diseased lung tissue?

The inhomogeneity of the lung is associated with inhomogeneous distribution of forces and obviously of mechanical power which might subsequent result in the prior lung dependent reason for the progression of VILI [40]. Indeed, differences in elasticity were suggested to concentrate the applied forces by doubling [41]. For a determined mechanical power, intensity of ventilation is increased in lung tissue with fewer ventilated areas and at the interface between lung areas with different mechanical properties [15, 42]. Apart from the dependence of the lung surface, VILI might also be impaired by the open/closed interfaces which were associated with increased [(18F)FDG] uptake and subsequent increased proportion of lung condition severity [40, 43]. Finally, this suggests the total area of the well-inflated lung tissue as well as the inhomogeneous poorly inflated or uninflated lung tissue important potential parameter for normalization of mechanical power [41]. Recently, Coppola et al. reported that normalization of mechanical power as well as respiratory system compliance were prior compared to the mere values in prediction of mortality [33]. Normalization was based on a whole lung CT scan at 5 cm H2O of PEEP and performed after a recruitment maneuver and lung gas volume and amount of well-inflated tissue were computed [33, 41].

Should ECMO be indicated to reduce intensity of ventilation?

Whether an earlier time point for initiation of ECMO therapy might an option, is currently under investigation (NCT04341285). Indeed, after implementation of the veno-venous EMCO therapy a so called “lung rest” with limited inspiratory plateau pressure (Pplat) < 25 cm H2O which is highly recommended by the ELSO is mostly feasible [6]. Further reductions in the Pplat below 20 cm H2O were reported to be associated with fewer VILI and improved patient outcome [45, 47]. Ultra-protective- or even near-apneic –ventilation during ECMO support was reported to attenuate lung injury due to decreased Vt and driving pressure [45, 48]. Concordantly, Araos et al. [49] reported, that near-apneic ventilation caused histologic less lung injury compared to both, a non-protective and conventional ventilatory strategy in an experimental ARDS ECMO pig model. However, others detected no superiority of ultra-protective ventilation strategies during vvECMO [46, 50]. Particularly, near apneic ventilation strategies are associated with a risk of atelectasis with subsequent secondary infections and severe induction of ventilation/perfusion mismatch unless PEEP is appropriately increased to keep part of the lung open [51]. Additionally, ultra-protective ventilation requires deep sedation which is necessary to avoid patient—ventilator asynchrony. Which ventilation mode and subsequent lung unloading is necessary to secure recovery, healing and repair need to be determined by clinical trials.

An international multicenter prospective cohort study revealed, that most high ECMO volume centers prefer a “lung rest” pressure-controlled lung protective ventilation strategy [46]. In contrast, pre-ECMO ventilation intensity was less considered and mechanical power prior vvECMO implementation was with 26.1 ± 12.7 J/min particularly high [46]. This might be explained, that EMCO therapy is complex, labor-intensive, expensive and moreover a highly invasive procedure. Therefore some centers apply ECMO therapy as a kind of “last rescue” procedure for a severely hypoxemic population after several trials of optimal conventional ventilation, prone positioning and neuromuscular blockers have failed. However, increased duration of mechanical ventilation before ECMO therapy initiation might be associated with higher mortality rates after ECMO therapy. Recent logistic regression analyses revealed greater delay from endotracheal intubation to ECMO initiation is independently associated with 6 month mortality [17]. Otherwise, the mere duration of ventilation prior to ECMO implantation was not associated with increased mortality other reports [4]. More important seems to be the intensity of ventilation and particularly even short durations of high intensity ventilation might cause lung injury [32]. Subsequently, an individualized comprehensive twice daily analysis of ventilation parameters particularly mechanical power during the pre-ECMO period is highly recommended. If sufficient oxygenation and/or decarboxylation cannot be achieved with at least moderate ventilation intensity, a MP and/or DP limited ventilation strategy with subsequent lung unload by vvECMO should be critically evaluated.

How could particularly MP be implemented during daily management?

The extended reference equation to calculate MP by Gattinoni et al. represents the most precise calculation [52, 53]. However, some variables like airway and tissue resistance or elastance of the respiratory system are complex to measure within the clinical routine setting [53]. Moreover, the application of this formula requires muscle relaxation and volume-controlled ventilation of the patients [53]. Meanwhile, several simplifications were developed for an application during the daily routine. For pressure-controlled ventilation two accurate equations were suggested, but require the knowledge of resistances and respiratory system compliance which are not determined within the daily routine [53,54,55]. However, Becher et al. recently suggested a simplified equation for MP calculation in pressure-controlled ventilated patients and Chiumello et al. reported bedside calculations of MP during volume- and pressure-controlled mechanical ventilation [55, 56]. Although this equations might be associated with a small bias of overestimation, it seems to be accurate and easy applicable during the daily routine [53]. Finally, the equations might also be applicable to spontaneous ventilation, but studies of accuracy and potential simplifications are currently lacking [57]. The current major limitation is, that airway pressure, flow and esophageal pressure are affected counter-directionally by the action of the ventilator and the respiratory muscles [57,58,59].

Another disadvantageous of MP is the requirement for normalization on well aerated lung tissue. However, most vvECMO centers conduct CT scans as standardized diagnostic procedure. Alternately, normalization might be based on the lung compliance with a decreased predictive performance compared to the well-aerated tissue and the acceptance of a potential collinearity between MP and compliance [33]. A MP and DP-based ventilation strategy is shown in Fig. 2. Extended and simplified MP equations for volume—and pressure-controlled ventilation are shown in Figs. 3 and 4.

Fig. 4
figure 4

Simplified equations for (1) pressure-controlled ventilation [75] and (2) volume-controlled ventilation [76]. Both equations were approved to approximate the surrogate formulas well enough for application during daily routine management. [56]

Which open questions need to be addressed in large clinical trials?

The relevance of DP and MP for vvECMO indication is just beginning. The key question clinical trials should address is the time point at which a lung unload and vvECMO support is necessary to avoid further ventilator-induced lung injury e.g., is the prize of sufficient oxygenation reasonable? With other words, threshold levels for pre-defined equations in dependence of ventilation strategy have to be determined. If adaptions of the ventilator strategy do not result in decreased MP, vvECMO implantation should be re-checked critically. However, the relative relevance of the MP components on VILI are currently not completely clear. Indeed, mathematically, MP increases with the exponent of 2 of Vt, the exponent of 1.4 of the RR and the exponent of 1 of PEEP. Whether an optimal composition of these parameters might be preferable to reduce VILI should be evaluated during the pre-ECMO ventilation strategy [60, 61].

However, in the case of ARDS, the development of innovative trial design is generally associated with several challenges. First, molecular biology of ARDS is very heterogeneous between patients with considerably differences in biomarkers of key pathways including inflammation, coagulation, and alveolar epithelial injury [62,63,64]. Therefore, a rigorously phenotyped patient large collective seems to be essential [62, 64]. Recently, the development of assay platforms based on protein-based enrichments strategy was suggested by Beitler et al. which might be most efficient to overcome this task [62, 65].

Second, outcome definition of ARDS in clinical trials is challenging [62]. Although mortality is frequently used in clinical trials as endpoint, ARDS-related risk of mortality differs considerably between patients and diseases. Moreover, endpoints other than mortality have to address death as a competing risk. However, this might be resolved by the application of a ranked composite score which compares each patient to all other patients encompassed in the study by vital status and subsequently, only if both patient in pair survive by the second to be analyzed outcome parameter. [62, 66, 67]

Third, VILI-related lung injury and effect of different ventilator strategies on VILI are not simply detected at bedside and separated form disease-related lung injury [68]. Frequently, global respiratory mechanics and the degree of ventilator support required were applied to characterize ARDS suffering patients [62]. However, particularly to separate disease from ventilator-related lung injury and to evaluate thresholds of DP and MP for ECMO indication, a standardized preclinical animal approach with an option of post-mortal lung injury characterization and associated biomarker and proteome analysis seems a reasonable approach. However, particularly lung compliance, airflow and VILI-induced inflammation seem to be highly species dependent [69]. In contrast to humans, most of the elastic recoil measured in intact mice can be attributed to the lung as chest wall and further thoracic structures are very compliant [69, 70]. Moreover, species dependent differences in inflammation and innate immunity were reported [71]. Exemplarily, Toll like receptor 4 from humans and mice recognize different lipopolysaccharide [72] and mice lack the CXCL8 gene coding IL8 [73, 74], a potent neutrophil chemotactic factor with a key role in the pathogenesis of ARDS. Although some this limitations might be overcome by the application of a large animal model, currently none of these models adequately reproduce the full characteristics of human ARDS. Therefore, subsequent bench to bedside approaches are indispensable.

Conclusion

The most challenging issues in ARDS patient treatment are the heterogeneity of the population and the continuously changing pulmonary circumstances. Therefore, an individualized patient and continuously adapted ventilation strategy is highly recommended (Fig. 2). Although much work has to be done to evaluate these strategies and thresholds, current evidence suggests, that driving pressure guided ventilation might decrease mechanical power and decreased mechanical power seems to be associated with decreased VILI. In particular, the mechanical power normalized to well inflated tissue and to respiratory system compliance were reported being independently associated with mortality of ARDS suffering patients [33]. The landmark paper by Umer et al. [32] analyzed 13,939 patients and suggested that higher intensities of mechanical ventilation reflected by driving pressure and mechanical power was harmful, even for short durations.  Whether finally limited exposure of driving pressure and mechanical power, which might subsequent result in earlier initiation of extracorporeal support (ECMO, ECCO2R) improve ARDS patients` outcome has to be investigate by randomized controlled trials. But until the results are available, we highly recommend to implement driving pressure and mechanical power during daily treatment of ARDS patients and subsequent ECMO therapy.

Availability of data and materials

Not applicable.

References

  1. Barbaro RP, MacLaren G, Boonstra PS, et al. Extracorporeal membrane oxygenation support in COVID-19: an international cohort study of the Extracorporeal Life Support Organization registry. Lancet. 2020;396:1071–8.

    Article  CAS  PubMed  PubMed Central  Google Scholar 

  2. Barbaro RP, MacLaren G, Boonstra PS, et al. Extracorporeal membrane oxygenation for COVID 19: evolving outcomes from the international Extracorporeal Life Support Organization Registry. Lancet. 2021;398:1230–8.

    Article  CAS  PubMed  PubMed Central  Google Scholar 

  3. Whebell S, Zhang J, Lewis R, et al. Survival benefit of extracorporeal membrane oxygenation in severe COVID-19: a multi-centre-matched cohort study. Intensive Care Med. 2022;48:467–78.

    Article  CAS  PubMed  PubMed Central  Google Scholar 

  4. Karagianndis C, Slutsky AS, Bein T, et al. Complete countrywide mortality in COVID patients receiving ECMO in Germany throughout the first three waves of pandemic. Crit Care. 2021;25:413.

    Article  Google Scholar 

  5. Herrmann J, Lotz C, Karagiannidis C, et al. Key characteristics impacting survival of COVID-19 extracorporeal membrane oxygenation. Crit Care. 2022;26:190.

    Article  PubMed  PubMed Central  Google Scholar 

  6. Combes A, Hajage D, Capellier G, et al. Extracorporeal membrane oxygenation for severe acute respiratory distress syndrome. N Engl J Med. 2018;378:1965–75.

    Article  PubMed  Google Scholar 

  7. Persico N, Guervilly C, Roch A, et al. Extracorporeal membrane oxygenation in acute respiratory distress syndrome: why is the EOLIA trial important? Ann Transl Med. 2018;6(Suppl 1):S20.

    Article  PubMed  PubMed Central  Google Scholar 

  8. Tonna JE, Abrams D, Brodie D, et al. Management of adult patients supported with venovenous extracorporeal membrane oxygenation (vvECMO): guideline from the extracorporeal life support organization (ELSO). ASAIO. 2021. https://doi.org/10.1097/MAT.0000000000001432.

    Article  Google Scholar 

  9. Dreyfuss D, Soler P, Basset G, et al. High inflation pressure pulmonoary edema: respective effects of high airway pressure, high tidal volume, and positive end-expiratory pressure. Am Rev Respir Dis. 1988;137:1159–64.

    Article  CAS  PubMed  Google Scholar 

  10. Slutsky AS, Ranieri VM. Ventilator-induced lung injury. N Engl J. 2013;369:22.

    Article  Google Scholar 

  11. Briel M, Meade M, Mercat A, et al. Higher vs. lower positive end-expiratory pressure in patients with acute lung injury and acute respiratory distress syndrome. JAMA. 2010;303:865–73.

    Article  CAS  PubMed  Google Scholar 

  12. Papazian L, Aubron C, Brochard L, et al. Fromal guidelines: management of acute respiratory distress syndrome. Ann Intensive Care. 2019;9:69.

    Article  PubMed  PubMed Central  Google Scholar 

  13. Bellani G, Laffey JG, Pharn T, et al. Epidemiology, patterns of care, and mortality for patients with acute respiratory distress syndrom in intensive care units in 50 countries. JAMA. 2016;315(8):788–800.

    Article  CAS  PubMed  Google Scholar 

  14. Cressoni M, Cadringher P, Chiurazzi C, et al. Lung inhomogeneity in patients with acute respiratory distress syndrome. Am J Respir Crit Care Med. 2014;189:149–58.

    Article  CAS  PubMed  Google Scholar 

  15. Mead J, Takishima T, Leith D. Stress distribution in lungs: a model of pulmonary elasticity. J Appl Physiol. 1970;28:596–608.

    Article  CAS  PubMed  Google Scholar 

  16. Budego G, Retamal J, Bruhn A. Driving pressure: a marker of severity, a safety limit, or a goal for mechanical ventilation. Crit Care. 2017;21:199.

    Article  Google Scholar 

  17. Grieco DL, Chen L, Dres M, et al. Should we use driving pressure to set tidal volume. Curr Opin Crit Care. 2017;23:38–44.

    Article  PubMed  Google Scholar 

  18. Chiu LC, Hu HC, Hung CY, et al. Dynamic driving pressure associated mortality in acute respiratory distress syndrome with extracorporeal membrane oxygenation. Ann Intensive Care. 2017;7:12.

    Article  PubMed  PubMed Central  Google Scholar 

  19. Haudebourg AF, Tuffet S, Perier F, et al. Driving pressure-guided ventilation decreases the mechanical power compared to predicted body weight-guided ventilation in the acute respiratory distress syndrome. Crit Care. 2022;26:185.

    Article  PubMed  PubMed Central  Google Scholar 

  20. Gattinoni L, Presenti A. The concept of “baby lung.” Intensive Care Med. 2005;31:776–84.

    Article  PubMed  Google Scholar 

  21. Gattinoni L, Presenti A, Avalli L, et al. Pressure-volume curve of total respiratory system in acute respiratory failure. Computed tomographic scan study. Am Rev Respir Dis. 1987;136:730–6.

    Article  CAS  PubMed  Google Scholar 

  22. Amato MBP, Meade MO, Slutsky AS, et al. Driving pressure and survival in the acute respiratory distress syndrome. N Engl Med. 2015;382:8.

    Google Scholar 

  23. Guerin C, Papazian L, Reignier J, et al. Effect of driving pressure on mortality in ARDS patients during lung protective mechanical ventilation in two randomized controlled trials. Crit Care. 2016;20:384.

    Article  PubMed  PubMed Central  Google Scholar 

  24. Talmor D, Sarge T, Malhotra A, et al. Mechanical ventilation guided by esophageal pressure in acute lung injury. N Engl J Med. 2008;359:2095–104.

    Article  CAS  PubMed  PubMed Central  Google Scholar 

  25. Gattinoni L, Marini JJ, Pesenti A, et al. The “baby lung” became an adult. Intensive Care med. 2016;42:663–73.

    Article  PubMed  Google Scholar 

  26. Goligher E, Costa ELV, Yarnell CJ, et al. Effect of lowering VT on mortality in acute respiratory distress syndrome varies with respiratory system Elastance. Am J Respir Crit Care. 2021;203:1378–85.

    Article  CAS  Google Scholar 

  27. Marini JJ, Rocco PRM, Gattinoni L. Static and dynamic contributors to VILI in clinical practice: pressure, energy, and power. Am J Respir Crit Care Med. 2020;201(7):767–74.

    Article  PubMed  PubMed Central  Google Scholar 

  28. Silva PL, Pelosi P, Rocco PR. Understanding the mysteries of mechanical power. Anesthesiology. 2020;132:949–50.

    Article  PubMed  Google Scholar 

  29. Collino F, Rapetti F, Vasques F, et al. Positive end-exspiratory pressure and mechanical power. Anesthesiology. 2019;130:119–30.

    Article  PubMed  Google Scholar 

  30. Sahetya SK, Mallow C, Sevransky JE, et al. Society of critical care medicine discovery network critical illness outcomes study investigators: association between hospital mortality and inspiratory airway pressure in mechanically ventilated patients without acute respiratory distress syndrome: a prospective cohort study. Crit Care. 2019;23:367.

    Article  PubMed  PubMed Central  Google Scholar 

  31. Neto AS, Hemmes SN, Barbas CS, et al. Association between driving pressure and development of postoperative pulmonary complications in patients undergoing mechanical ventilation for general anaesthesia: a meta analysis of individual patient data. Lancet Respir Med. 2016;4:272–80.

    Article  PubMed  Google Scholar 

  32. Umer M, Jüni P, Hansen B, et al. Time-varying intensity of mechanical ventilation and mortality in patients with acute respiratory failure: a registry-based, prospective cohort study. Lancet Respir Med. 2020;8:905–13.

    Article  Google Scholar 

  33. Coppola S, Cacciopoola A, Froio S, et al. Effect of mechanical power on intensive care mortality in ARDS patients. Crit Care. 2020;24:246.

    Article  PubMed  PubMed Central  Google Scholar 

  34. Santos RS, de Maia LD, Iliveria MV, et al. Biologic impact of mechanical power at high and low tidal volumes in experimental mild acute respiratory distress syndrome. Anesthesiology. 2018;128:1193–206.

    Article  PubMed  Google Scholar 

  35. Serpy Neto A, Deliberato RO, Johnson AEW, et al. Mechanical power of ventilation is associated with mortality in critically ill patients: an analysis of patients in two observational cohorts. Intensive Care Med. 2018;44:1914–22.

    Article  Google Scholar 

  36. Mailo G, Collino F, Vasques F, et al. Reclassifying acute respiratory distress syndrome. Am J Respir and Crit Care Med. 2018;197:1586–95.

    Article  Google Scholar 

  37. Cavalacanti AB, Suzumura EA, Laranjeira LN, et al. Writing Group for the Aleveolar recruitment for acute respiratory distress syndrome Trial (ART) Investigators. Effect of lung recruitment and titrated positive end-exspiratory pressure (PEEP) vs low PEEP on mortality in patients with acute respiratory distress syndrome: a randomized clinical trial. JAMA. 2017;318:1335–45.

    Article  Google Scholar 

  38. Vassalli F, Pasticci I, Romitti F, et al. Does iso-mechanical power lead to iso-lung damage? An experimental study in a porcine model. Anesthesiology. 2020;132:1126–37.

    Article  CAS  PubMed  Google Scholar 

  39. Cressoni M, Gotti M, Chiurazzi C, et al. Mechanical power and development of ventilator-induced lung injury. Anesthesiology. 2016;124:1100–8.

    Article  PubMed  Google Scholar 

  40. Gattiononi L, Tonetti T, Cressoni M, et al. Ventilator-related causes of lung injury: the mechanical power. Intensive care med. 2026;42:1567–75.

    Article  Google Scholar 

  41. Cressoni M, et al. Lung inhomogeneity in patients with acute respiratory distress syndrome. Am J Respir Crit Care Med. 2014;189:149–58.

    Article  CAS  PubMed  Google Scholar 

  42. Silva PL, Ball L, Rocco PRM, et al. Power to mechanical power to minimize ventilator-induced lung injury? Intensive Care Med. 2019;7:38.

    Article  Google Scholar 

  43. Cressoni M, Chimello D, Chiurazzi C, et al. Lung inhomogeneities inflation and [18F]2-fluro-2-deoxy-D-glucose uptake rate in acute respiratory distress syndrome. Eur Respir J. 2016;47:233–42.

    Article  CAS  PubMed  Google Scholar 

  44. Gattioni L, Caironi P, Pelosi P, et al. What has computed tomography taught us about the acute respiratory distress syndrome? Am J Respir Crit Care Med. 2001;164:1701–11.

    Article  Google Scholar 

  45. Rozencwajg S, Guihot A, Franchineau G, et al. Ultra-protective ventilation reduces biotrauma in patients on venovenous extracorporeal membrane oxygenation for severe acute respiratory distress syndrome. Crit Care Med. 2019;47:1505–12.

    Article  PubMed  Google Scholar 

  46. Schmidt M, Pham T, Arcadipane A, et al. Mechanical ventilation management during extracorporeal membrane oxygenation for acute respiratory distress syndrome. An international multicenter prospective cohort. Am J Respir Crit Care Med. 2019;200:1002–12.

    Article  CAS  PubMed  Google Scholar 

  47. Quintel M, Busana M, Gattioni L. Breathing and ventilation during extracorporeal membrane oxygenation: how to find the balance between rest and load. Am J Respir Crit Care Med. 2019;200:954–6.

    Article  PubMed  PubMed Central  Google Scholar 

  48. Graf PT, Boesing C, Brumm I, et al. Ultraprotective versus apneic ventilation in acute respiratory distress syndrome patients with extracorporeal membrane oxygenation: a physiological study. J Int Care. 2022;10:12.

    Google Scholar 

  49. Araos J, Alegria L, Garcia P, et al. Near-apneic ventilation decreases lung injury and fibroproliferation in an acute respiratory distress syndrome model with extracorporeal membrane oxygenation. Am J Respir Crit Care Med. 2019;199:603–12.

    Article  CAS  PubMed  Google Scholar 

  50. Guervilly C, Fournier T, Chommeloux J, et al. Ultra-lung-protective ventilation and biotrauma in severe ARDS patients on veno-venous extracorporeal membrane oxygenation: a randomized controlled study. Crit Care. 2022;26:383.

    Article  PubMed  PubMed Central  Google Scholar 

  51. Schmidt M, Pellegrino V, Combes A, et al. Mechanical ventilation during extracoprporeal membrane oxygenation. Crit Care. 2014;18:203.

    Article  PubMed  PubMed Central  Google Scholar 

  52. Gattinoni L, Tonetti T, Cressoni M, et al. Ventilator-related causes of lung injury: the mechanical power. Intensive Care Med. 2016;42:1564–75.

    Article  Google Scholar 

  53. Goisa L, Busana M, Paticci I, et al. Mechanical power at a glance3: a simple surrogate for volume controlled ventilation. Intensive Care Med Exp. 2019;7:61.

    Article  Google Scholar 

  54. de Meijden S, Molenaar M, Somhorst P, et al. Calculating mechanical power for pressure-controlled ventilation. Intensive Care Med. 2019;45:1495–7.

    Article  PubMed  Google Scholar 

  55. Becher T, van der Staay M, Schadler D, et al. Calculation of mechanical power for pressure controlled ventilation. Intensive Care Med. 2019;45:1321–3.

    Article  PubMed  Google Scholar 

  56. Chiumello D, Gotti M, Guanziroli M, et al. Bedside caluculation of mechanical power during volume- and pressure-controlled mechanical ventilation. Crit Care. 2020;24:417.

    Article  PubMed  PubMed Central  Google Scholar 

  57. Huhle R, Serpa Neto A, Schlutz MJ, et al. Is mechanical power the final word on ventilator-induced lung injury?-No. Ann Transl Med. 2018;6:394.

    Article  PubMed  PubMed Central  Google Scholar 

  58. Cressoni M, Cadringher P, Chiurazzi C, et al. Lung inhomogeneity in patients with acute respiratory distress syndrome. Am J Respir Crit Care Med. 2014;189:149–58.

    Article  CAS  PubMed  Google Scholar 

  59. Marini JJ, Jaber S. Dynamic predictors of VILI risk: beyond the driving pressure. Intensive Care Med. 2016;42:1597–600.

    Article  PubMed  Google Scholar 

  60. Gattinoni L, Tonetti T, Cressoni M, et al. Ventilator related causes of lung injury: the mechanical power. Intensvie Care Med. 2016;42:1567–75.

    Article  CAS  Google Scholar 

  61. Vasques F, Duscio E, Pasticci I, et al. Is the mechanical power the final word on ventilator-induced lung injury?—We are not sure. Ann Transl Med. 2018;6:395.

    Article  PubMed  PubMed Central  Google Scholar 

  62. Beitler JR, Thompson TB, Baron RM, et al. Advancing precision medicine for acute respiratory distress syndrome. Lancet Respir Med. 2022;10:107–20.

    Article  CAS  PubMed  Google Scholar 

  63. Meyer NJ, Calfee C. Novel translational approaches to the search of precision therapies for acute respiratory distress syndrome. Lancet Respir Med. 2017;5:512–23.

    Article  PubMed  PubMed Central  Google Scholar 

  64. Wick KD, McAuley DF, Levitt JE, et al. Promises and challenges of personalized medicine to guide ARDS therapy. Crit Care. 2021;25:404.

    Article  PubMed  PubMed Central  Google Scholar 

  65. Matthay MA, Arabi YM, Siegel ER, et al. Phenotypes and personalized medicine in the acute respirtory distress syndrome. Intensive Care Med. 2020;46:2136–52.

    Article  CAS  PubMed  PubMed Central  Google Scholar 

  66. Beitler JR, Sarge T, Banner-Goodspeech VM, et al. Effect of titrating positive end-exspiratory pressure (PEEP) with an eosophageal pressure guided strategy vs. an empirical high PEEP-FiO2 strategy on death and days free form mechanical ventilation among patients with acute respiratory distress syndrome: a randomized clinical trial. JAMA. 2019;321:846–57.

    Article  PubMed  PubMed Central  Google Scholar 

  67. Novack V, Beitler JR, Yisshak-Sade M, et al. Alive and ventilator free: a hierarchical, composite outcome for clinical trials in acute respiratory distress syndrome. Crit Care Med. 2020;48:158–66.

    Article  PubMed  PubMed Central  Google Scholar 

  68. Beitler JR. Bedside respiratory physiology to detect risk of lung injury in acute respiratory distress syndrome. Curr Opin Crit Care. 2019;25:3–11.

    Article  PubMed  PubMed Central  Google Scholar 

  69. Irvin IG, Bates JHT. Measuring the lung function in the mouse: the challenge of size. Respir Res. 2003;4:1–9.

    Article  Google Scholar 

  70. Takezawa CG, Rabold R, Mitzner W. Differenztial lung mechanics are genetically determined in inbred murine strains. J Appl Physiol. 1999;86:1764–9.

    Article  Google Scholar 

  71. Matute-Bello G, Frevert CW, Martin TR. Animal models of acute lung injury. Am J Physiol Lung Cell Mol Physiol. 2008;295:L3379–99.

    Google Scholar 

  72. Hajjar AM, Ernst RK, Tsai JH, et al. Human Toll like receptor 4 recognizes host-specific LPS modfications. Nat Immunol. 2002;3:354–9.

    Article  CAS  PubMed  Google Scholar 

  73. Baughman RP, Gunther KL, Rashkin MC, et al. Changes in the inflammatory response of the lung during acute respiratory distress syndrome: prognostic indicators. Am J Respir Crit Care Med. 1996;154:76–81.

    Article  CAS  PubMed  Google Scholar 

  74. Miller EJ, Cohen AB, Matthay M. Increased interleukin 8 concentations in the pulmonary edema fluid of patients with acute respiratory distress syndrome from sepsis. Crit Care Med. 1996;24:1448–54.

    Article  CAS  PubMed  Google Scholar 

  75. Goisa L, Busana M, Pasticci L, et al. Mechanical power at a glance: a simple surrogate for volume-controlled ventilation. Intensive Care Med Exp. 2019;7:61.

    Article  Google Scholar 

  76. Becher T, van der Staay M, Schadler D, et al. Calculation of mechanical power for pressure-controlled ventilation. Intensive Care Med. 2019;45:1321–3.

    Article  PubMed  Google Scholar 

Download references

Funding

This publication was supported by the Open Access Publication Fund (Projekt DEAL) of the University Wuerzburg.

Author information

Authors and Affiliations

Authors

Contributions

KH and TR contributed to the conception of the work, KH and EK drafted the manuscript, TR designed figures, PM and TR critically revised the work. All authors read and approved the final version of the manuscript.

Corresponding author

Correspondence to K. Hoppe.

Ethics declarations

Ethical approval and consent to participate

Not applicable.

Competing interests

The authors declare no competing interests.

Additional information

Publisher's Note

Springer Nature remains neutral with regard to jurisdictional claims in published maps and institutional affiliations.

Rights and permissions

Open Access This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.

Reprints and Permissions

About this article

Verify currency and authenticity via CrossMark

Cite this article

Hoppe, K., Khan, E., Meybohm, P. et al. Mechanical power of ventilation and driving pressure: two undervalued parameters for pre extracorporeal membrane oxygenation ventilation and during daily management?. Crit Care 27, 111 (2023). https://doi.org/10.1186/s13054-023-04375-z

Download citation

  • Received:

  • Accepted:

  • Published:

  • DOI: https://doi.org/10.1186/s13054-023-04375-z

Keywords

  • ARDS
  • Ventilation
  • ECMO indication
  • Mechanical power
  • Driving pressure