Wearable ultrasound and provocative hemodynamics: a view of the future

Background

A new technology with exciting implications for critical care medicine was recently described by Wang and colleagues [1]. They report a bio-adhesive ultrasound (BAUS) with robust skin-coupling, comfort and excellent acoustic properties. In their feasibility study, they describe a small ultrasound transducer adhering over multiple tissues and organ systems, capable of acquiring 2-dimensional, brightness mode (B-mode) images in addition to pulsed wave and color flow Doppler. Furthermore, they reference a separate, fully contained, wireless, wearable, ultrasound system developed by our group [2]. We believe that both of these technologies may provide powerful, easy-to-use, hemodynamic measures for the twenty-first century intensivist [3]. In this Comment article, we briefly describe some significant physiological, clinical and practical implications of wearable ultrasound technology in the intensive care unit (ICU).

Main text

Physiological implications

Wang and colleagues characterize simultaneous common carotid artery and internal jugular vein B-mode images with the BAUS [1]. Specifically, they show that when a healthy subject shifts from upright-to-supine positions whilst wearing the BAUS, the jugular vein changes shape—from a collapsed to a distended profile. As Wang and colleagues note, this could act as a surrogate for changing right atrial pressure. Separately, the authors capture quantitative spectral Doppler ultrasound of the common carotid artery and its change from resting to post-exercise, demonstrating that the carotid artery velocity time integral (VTI) increases with exercise-induced cardiac output augmentation.

Though not explicitly described by Wang and colleagues, we believe what is most exciting about these data is simultaneous, noninvasive venous and arterial ultrasound surrogates for cardiac input and output, respectively. This novel paradigm has substantial inpatient and outpatient diagnostic and therapeutic implications—for example, real-time inferences of the Frank–Starling mechanism [2, 4,5,6]. This is of particular interest in the ICU. More specifically, during a provocative maneuver (e.g., passive leg raise (PLR) or rapid fluid challenge), increasing right atrial pressure coincident with little change in arterial flow intimates lost preload reserve [4], a feature typical of congestive heart failure [7] and septic shock [8]. While experts recommend confirming that right atrial pressure rises with a PLR to ensure that venous return challenges the Frank–Starling mechanism [9], this is cumbersome in practice. With wearable ultrasound technology, tracking these hemodynamic measures in real time, and in response to provocative maneuvers, is of tremendous utility to intensivists [10].

Clinical implications

A recently published framework speculates how simultaneously acquired venous and arterial Doppler ultrasound could inform patient therapy and posits that Doppler ultrasound accompany all advanced critical care echocardiography to better delineate unique hemodynamic phenotypes [6]. For instance, dynamic fluid intolerance describes a hypo-perfused patient subtype with suggestive signs of low filling pressure (e.g., flat jugular vein, collapsing inferior vena cava, low central venous pressure) who is, nevertheless, found to be fluid unresponsive during a dynamic maneuver such as a PLR [6]. We envision how this phenotype could be detected using the ultrasound system described by Wang and colleagues. A BAUS worn over the apex of the left ventricle with a pulsed wave Doppler gate placed at the left ventricular outflow tract (LVOT) measures LVOT VTI in the baseline, semi-Fowler position; a separate BAUS on the neck simultaneously images a collapsed jugular vein. Then on PLR, the jugular vein is observed to distend (i.e., augmented right atrial pressure), while the LVOT VTI does not rise (i.e., stroke volume is unchanged). Thus, this patient who might have been mistakenly labeled as ‘fluid responsive’ or ‘fluid tolerant’ [11] based on signs of venous filling, is accurately determined to be ‘fluid unresponsive’ or ‘fluid intolerant’ following the provocative PLR maneuver. We have observed a similar phenotype using a wearable, continuous wave Doppler ultrasound system that records internal jugular and common carotid artery Doppler spectra as surrogates for changing right atrial pressure and stroke volume, respectively [4, 12] (Fig. 1).

Fig. 1
figure 1

Wearable ultrasound and provocative hemodynamics. A Picture of a commercial wireless, wearable Doppler ultrasound system on a volunteer. B Graphical illustration of wearable ultrasound beam with simultaneous insonation of the common carotid artery and internal jugular vein. C Physiological framework showing how simultaneous venous and arterial ultrasound tracks the two axes of the Frank–Starling curve. Venous ultrasound tracks changing cardiac ‘input’ (e.g., right atrial pressure) and arterial ultrasound follows cardiac ‘output’ (e.g., stroke volume). Moving from point iii. to iv. illustrates a patient with low filling pressure who is fluid unresponsive [4]. D Simultaneous common carotid and internal jugular Doppler of a healthy volunteer in the semi-recumbent position. The jugular spectrum is non-pulsatile and higher velocity, suggesting a collapsed vein (i.e., low right atrial pressure). E Simultaneous common carotid and internal jugular Doppler of the same volunteer during PLR. The jugular velocity falls and becomes pulsatile, following the right atrial pressure waveform (i.e., the Doppler systolic ‘s’ wave is formed by the x-descent, and the diastolic ‘d’ wave is formed by the ‘y’ descent); this change is consistent with an enlarging jugular vein and right atrial pressure, and concomitantly, the carotid artery velocity time integral rises. This physiology is most compatible with moving from point i. to ii. in C

Practical implications

Wang and colleagues describe impressive coupling properties of their wearable BAUS. From a practical standpoint, we have observed that there needs to be a very good balance between firm skin adhesion and the ability to reposition. This is particularly important during signal acquisition because obtaining a good ultrasound window can be challenging and, in certain patient populations, requires multiple attempts to optimize signal. To account for these practical challenges, the fully contained, wireless, wearable Doppler ultrasound system described by our group is coupled to the skin using off-the-shelf ultrasound gel [2]. The ultrasound beam dimensions and properties were chosen specifically to favor simultaneous acquisition of Doppler spectra from the common carotid artery and internal jugular vein and not for anatomical flexibility, which can come at the cost of achieving an optimal signal. The device is affixed to the neck using a variety of adhesives, but when sealed, can maintain intermittent Doppler ultrasound scanning for multiple days and across a range of human activities.

Conclusion

The recent work of Wang et al. offers a view of the future [1]. We expect that point-of-care ultrasound in the twenty-first century ICU will continue to miniaturize, become wearable and free itself from unwieldy tethers [13]. The new technologies described above could simultaneously and noninvasively acquire and transmit surrogates for cardiac input and output; we propose that this generates the slope of the cardiac function curve in real time. This paradigm could be an especially important supplement to functional hemodynamic monitoring [14] and an additional step toward personalized therapy in the ICU.

Availability of data and materials

Not applicable.

Abbreviations

B-mode:

Brightness mode

BAUS:

Bio-adhesive ultrasound

ICU:

Intensive care unit

VTI:

Velocity time integral

PLR:

Passive leg raise

LVOT:

Left ventricular outflow tract

References

  1. Wang C, Chen X, Wang L, Makihata M, Liu H-C, Zhou T, Zhao X. Bioadhesive ultrasound for long-term continuous imaging of diverse organs. Science. 2022;377(6605):517–23.

    Article  CAS  Google Scholar 

  2. Kenny J-ÉS, Munding CE, Eibl JK, Eibl AM, Long BF, Boyes A, Yin J, Verrecchia P, Parrotta M, Gatzke R. A novel, hands-free ultrasound patch for continuous monitoring of quantitative Doppler in the carotid artery. Sci Rep. 2021;11(1):1–11.

    Article  Google Scholar 

  3. Barjaktarevic I, Kenny JÉS, Berlin D, Cannesson M. The evolution of ultrasound in critical care: from procedural guidance to hemodynamic monitor. J Ultrasound Med. 2021;40(2):401.

    Article  Google Scholar 

  4. Kenny J-ÉS, Barjaktarevic I, Mackenzie DC, Rola P, Haycock K, Eibl AM, Eibl JK. Inferring the Frank–Starling curve from simultaneous venous and arterial Doppler: measurements from a wireless, wearable ultrasound patch. Front Med Technol. 2021;3(16):676995.

    Article  Google Scholar 

  5. Kenny J-ES, Eibl JK, Mackenzie DC, Barjaktarevic I. Guidance of intravenous fluid by ultrasound will improve with technology. Chest. 2021;161(2):132–3.

    Article  Google Scholar 

  6. Kenny JES. Assessing fluid intolerance with Doppler ultrasonography: a physiological framework. Med Sci. 2022;10(1):12.

    Google Scholar 

  7. Abe Y, Akamatsu K, Furukawa A, Ito K, Matsumura Y, Haze K, Naruko T, Yoshiyama M, Yoshikawa J. Pre-load-induced changes in forward LV Stroke and functional mitral regurgitation: echocardiographic detection of the descending limb of Starling’s curve. JACC Cardiovasc Imaging. 2017;10(6):611–8.

    Article  Google Scholar 

  8. Monnet X, Shi R, Teboul J-L. Prediction of fluid responsiveness. What’s new? Ann Intensive Care. 2022;12(1):46.

    Article  Google Scholar 

  9. Monnet X, Teboul J-L. Passive leg raising. Intensive Care Med. 2008;34(4):659–63.

    Article  Google Scholar 

  10. Kenny J-ÉS. Functional hemodynamic monitoring with a wireless ultrasound patch. J Cardiothorac Vasc Anesth. 2021;35(5):1509–15.

    Article  Google Scholar 

  11. Kattan E, Castro R, Miralles-Aguiar F, Hernández G, Rola P. The emerging concept of fluid tolerance: a position paper. J Crit Care. 2022;71: 154070.

    Article  Google Scholar 

  12. Kenny J-ÉS, Barjaktarevic I, Mackenzie DC, Elfarnawany M, Yang Z, Eibl AM, Eibl JK, Kim C-H, Johnson BD. Carotid Doppler ultrasonography correlates with stroke volume in a human model of hypovolaemia and resuscitation: analysis of 48 570 cardiac cycles. Br J Anaesth. 2021;127(2):e60–3.

    Article  Google Scholar 

  13. Michard F, Pinsky MR, Vincent JL. Intensive care medicine in 2050: NEWS for hemodynamic monitoring. Intensive Care Med. 2017;43(3):440–2.

    Article  Google Scholar 

  14. Pinsky MR. Functional hemodynamic monitoring: current concepts in critical care. Curr Opin Crit Care. 2014;20(3):288.

    Article  Google Scholar 

Download references

Acknowledgements

Kyle Fredericks for image artwork.

Funding

No specific funding supported this work.

Author information

Authors and Affiliations

Authors

Contributions

JESK contributed to conception, drafting and revisions; CEM was involved in drafting and critical revisions; AME contributed to drafting and critical revisions; and JKE was involved in drafting and critical revisions. All authors read and approved the final manscript.

Corresponding author

Correspondence to Jon-Émile S. Kenny.

Ethics declarations

Ethics approval and consent to participate

Not applicable.

Consent for publication

Not applicable.

Competing interests

JESK, CEM, AME and JKE are employees of Flosonics Medical, a start-up working to commercialize a wearable Doppler ultrasound.

Additional information

Publisher's Note

Springer Nature remains neutral with regard to jurisdictional claims in published maps and institutional affiliations.

Rights and permissions

Open Access This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.

Reprints and permissions

About this article

Check for updates. Verify currency and authenticity via CrossMark

Cite this article

Kenny, JÉ.S., Munding, C.E., Eibl, A.M. et al. Wearable ultrasound and provocative hemodynamics: a view of the future. Crit Care 26, 329 (2022). https://doi.org/10.1186/s13054-022-04206-7

Download citation

  • Received:

  • Accepted:

  • Published:

  • DOI: https://doi.org/10.1186/s13054-022-04206-7

Keywords