Flow Index accurately identifies breaths with low or high inspiratory effort during pressure support ventilation

Abstract

Background

Flow Index, a numerical expression of the shape of the inspiratory flow-time waveform recorded during pressure support ventilation, is associated with patient inspiratory effort. The aim of this study was to assess the accuracy of Flow Index in detecting high or low inspiratory effort during pressure support ventilation and to establish cutoff values for the Flow index to identify these conditions. The secondary aim was to compare the performance of Flow index,of breathing pattern parameters and of airway occlusion pressure (P0.1) in detecting high or low inspiratory effort during pressure support ventilation.

Methods

Data from 24 subjects was included in the analysis, accounting for a total of 702 breaths. Breaths with high inspiratory effort were defined by a pressure developed by inspiratory muscles (Pmusc) greater than 10 cmH2O while breaths with low inspiratory effort were defined by a Pmusc lower than 5 cmH2O. The areas under the receiver operating characteristic curves of Flow Index and respiratory rate, tidal volume,respiratory rate over tidal volume and P0.1 were analyzed and compared to identify breaths with low or high inspiratory effort.

Results

Pmusc, P0.1, Pressure Time Product and Flow Index differed between breaths with high, low and intermediate inspiratory effort, while RR, RR/VT and VT/kg of IBW did not differ in a statistically significant way. A Flow index higher than 4.5 identified breaths with high inspiratory effort [AUC 0.89 (CI 95% 0.85–0.93)], a Flow Index lower than 2.6 identified breaths with low inspiratory effort [AUC 0.80 (CI 95% 0.76–0.83)].

Conclusions

Flow Index is accurate in detecting high and low spontaneous inspiratory effort during pressure support ventilation.

Background

Inspiratory support should maintain the inspiratory effort into a physiological range, sustainable by the patient. Under-assistance can lead to excessive inspiratory effort, generating a potentially harmful transpulmonary pressure leading to regional lung stress [1] and myotrauma [2] whereas over-assistance is associated with diaphragmatic atrophy and dysfunction [3].

Nonetheless, assessing inspiratory effort during assisted mechanical ventilation remains a clinical challenge, since a validated and affordable method to quantitatively assess it at the bedside is not yet available [4]. Monitoring the esophageal pressure (Pes), which is the gold standard to evaluate the pressure developed by the respiratory muscles (Pmusc), is relatively invasive and requires considerable technical expertise in order to correctly interpret the Pes waveforms [5, 6]. Breathing pattern parameters, such as tidal volume and respiratory rate, or the rapid shallow breathing index (RR/VT) [7] are used as surrogates to infer patient effort, but they may be inaccurate and misleading [8]. Other proposed measures, such as the airway occlusion pressure (P0.1) and the swing in airway pressure generated by respiratory muscle effort recorded during a brief airway opening occlusion at end-expiration (ΔPocc) are both affected by technical or conceptual limitations [9, 10]. In particular, P0.1 measured by mechanical ventilators has a measurement error of ± 2 cmH2O [10, 11]. Since the threshold for identifying excessive inspiratory effort is 3.5–4 cmH2O and 1 cmH2O for low inspiratory effort, this approximation is far from negligible. Moreover, in subjects with respiratory muscle weakness, P0.1 could be low even in the presence of a high effort and of insufficient inspiratory support. Concerning ΔPocc, it is useful only in evaluating high patient effort, it requires active intervention from the attending physician, it is not continuous and not all ventilators allow to perform occlusions during PSV [12, 13].

Recently, we demonstrated that the Flow Index, a numerical expression of the shape of the inspiratory flow-time waveform recorded during pressure support ventilation (PSV), is independently associated with patient effort [14]. The aim of the present study was to assess the accuracy of the Flow Index in detecting high or low inspiratory effort during PSV and to establish cutoff values for the Flow index to identify these conditions. As a secondary aim, we sought to compare the performance of the Flow index,of the breathing pattern parameters (RR, VT,and RR/VT) and of P0.1 in estimating inspiratory effort during PSV.

Methods

This study analyzed data collected during the Flow Index study [14] and was approved by the local ethical committee (Comitato Etico della Provincia di Brescia, NP4622).

Data was collected in patients admitted to the Intensive Care Unit (ICU) of Fondazione Poliambulanza, Brescia, Italy, who met all of the following criteria: age > 18 years, dependence on invasive mechanical ventilation (i.e. not ready to wean or having failed a spontaneous breathing trial on the day of the study [15]), being in PSV, having an esophageal balloon catheter already in place. Exclusion criteria were: mean arterial pressure < 60 mmHg, systolic arterial pressure > 180 mmHg, heart rate < 40 min−1 or > 150 min−1, PaO2/FIO2 < 150 mmHg, pH < 7.35 with PaCO2 > 45 mmHg, diagnosis of head injury, intracranial hemorrhage or cerebral ischemia.

A detailed description of the study protocol and measurements, and of the Flow index derivation formulas, is available elsewhere [14]. The mechanical ventilators in use for the study were Maquet Servo-i (Solna, Sweden) and GE-Datex Ohmeda S/5 Engstrom (Helsinki, Finland).

Protocol

In order to explore the whole spectrum of patient effort, three pressure support (PS) levels were applied to each patient: (1) the PS level at enrollment was defined as basal, (2) the lowest tolerated PS level without dyspnea while keeping the ratio between respiratory rate and tidal volume (RR/VT) < 100 min−1 L−1 was defined as low, and (3) the maximal tolerated PS level to achieve near relaxation was defined as high. The high PS level was obtained by progressively increasing the PS until all signs of inspiratory muscle activity disappeared after inspiratory triggering, as assessed by visual inspection of the waveform of Pes, airway opening pressure (Paw) and airflow. In order to avoid lung injury, the peak airway pressure was limited to a maximum of 30 cmH2O, regardless of achieving complete absence of inspiratory muscle activity. The three levels of PS were randomly applied for 20 min each, and all the remaining ventilatory variables (FiO2, inspiratory trigger, expiratory trigger) remained constant throughout the study, as previously set by the attending physician.

Measurements

At the end of each 20 min period of stable PS level, Paw at the ventilator Y connector, Pes, inspiratory and expiratory flow, VT and RR were recorded for 5 min (Datex-Ohmeda S/5 Collect; Datex-Ohmeda Division, Instrumentarium Corp., Helsinki, Finland). The sampling rate was 100 Hz.

Pes was measured by an esophageal balloon catheter (Marquat Gbm, Boissy-St-Léger Cedex, France) connected to a pressure transducer (AS3/CS3; Datex-Engstrom Division, Instrumentarium Corp., Helsinki, Finland).

The static recoil pressure of the chest wall (Pcw) was calculated as the product of the VT and the measured chest wall elastance (Ecw). Ecw was obtained as the ratio between the inspiratory change in Pes [end-inspiratory plateau esophageal pressure (Pplat,es) minus end-expiratory plateau esophageal pressure (Pexp,es)] and VT obtained during in a condition of near relaxation. The pressure generated by inspiratory muscles (Pmusc) was calculated as the maximal difference between Pcw and Pes (Fig. 1).

Fig. 1
figure 1

Airway pressure (Paw), esophageal pressure (Pes) and airflow curves from a study participant. Upper panel: change in esophageal pressure (Pes, continuous line) and in chest wall elastic recoil pressure (Pcw, dashed line). The maximum pressure generated by respiratory muscles (Pmusc) is the maximal distance between Pcw and Pes (double arrow line). Lower panel: flow trace

P0.1 was defined as the drop in esophageal pressure in the first 100 ms after the beginning of inspiration.

The Pressure–Time Product (PTP) was calculated as the area between Pcw and Pes during inspiratory flow multiplied by the respiratory rate, while PTPtot was calculated by adding the area included between Pcw starting from the beginning of inspiratory effort and Pcw starting from beginning of inspiratory flow (grey area on Fig. 1) to PTP, multiplied by the respiratory rate. Lung compliance and resistance were also calculated using the Least Squares Fitting Method on transpulmonary pressure [16]. Respiratory system compliance and resistance were then calculated with the Least Squares Fitting Method on airway pressure [17] during low inspiratory effort (defined as P0.1 < 1.7 cmH2O [17] and PTP < 50 cmH2O s−1 min−1 [10]).

Paw, Pes and flow traces were independently reviewed by two authors (FA and GN) and patients were excluded if the data were not reliable or if evident artefacts were present. All consecutive breaths obtained from the longest portion of the esophageal pressure waveform without swallowing artifacts (detected by a transient, sudden increase on the pressure trace) were used for the analyses.

The Flow Index

The detailed calculation of the Flow Index has been previously described [14]. Briefly, the portion of the inspiratory flow-time waveform included between the end of the ramp and before the expiratory trigger was fitted with the non-linear equation:

$$\dot{V} = a + b \cdot \Delta \,{\text{time}}^{{\text{c}}}$$
(1)

where the inspiratory flow (\(\dot{V}\)) is a function of time, of peak flow (a), of the rate of flow reduction (b) and of parameter c, which describes the downward facing concavity of the portion of the inspiratory flow waveform. The parameter c, calculated for every breath, was named Flow Index. The Flow Index describes the concavity of the curve using the same equation that computes the well-known Stress Index, which is calculated on airway pressure instead of inspiratory flow [18]. The Flow Index is equal to 1 when the inspiratory flow decreases linearly. If the waveform has an upward facing concavity, the Flow Index is < 1, whereas if the curve has a downward facing concavity, the Flow Index is > 1 (Fig. 2).

Fig. 2
figure 2

Procedure used to calculate Flow Index from the descending inspiratory portion of the flow waveform. Flow waveforms at the three different pressure support levels. The grey circles indicate the sampled inspiratory flow values, while the vertical lines indicate the cutting points used to select the descending inspiratory portion of the flow waveform. The red line shows the fitted model, calculated using Eq. 1 (detailed explanation in text). Abbreviations PSmin, minimum pressure support; PSbase, baseline pressure support; PSmax, maximum pressure support

Study outcome

The primary outcome was to validate the ability of the Flow Index in identifying low and high spontaneous inspiratory effort during PSV. In agreement with previously proposed cutoffs [6, 19, 20], breaths with high inspiratory effort were defined by a Pmusc greater than 10 cmH2O while breaths with low inspiratory effort were defined by a Pmusc lower than 5 cmH2O. As a secondary outcome, Flow Index was compared to other breathing pattern parameters used to monitor respiratory effort (RR, VT,and RR/VT) and to P0.1. In order to do so, we compared the areas under the receiver operating characteristic curves (AUC-ROC) of Flow Index and RR, VT, RR/VT and P0.1 to identify breaths with low or high inspiratory effort.

Statistical analysis

For the sample size calculation, we estimated a difference of 0.1 between the AUC of the Flow Index and the AUC of the RR/VT in detecting breaths at low inspiratory effort. Considering a type one error rate set at 0.05, we calculated that 670 breaths would be necessary to obtain a power of 0.9, considering a ratio between cases and controls of 0.2 [21].

Continuous variables were described with mean (standard deviation) or median [1st-3rd quartile] and analyzed with the Student's T-test or Mann–Whitney test in accordance with their distribution. The categorical variables were expressed as counts (%) and analyzed with the Fisher exact test.

Logistic mixed effect models were used to obtain a probability of the occurrence of high or low effort while taking into account the aggregate structure of the data. Individual patients were entered as random-effects in these models and the fixed-effect used for each different model was the variable under analysis (Flow Index, RR, VT/kg of ideal body weight [IBW], RR/VT). After fitting the models, the conditional prediction was used to assess the AUC of the ROC curve and estimate sensibility, specificity, positive predictive value (PPV) and negative predictive value (NPV). Confidence intervals at 95% (CI 95%) were obtained with 2000 bootstrapped samples [22].

To evaluate the utility of other indices used in clinical practice, the same procedure was used to estimate the AUC of ROC curves of RR, VT, RR/VT and P0.1 and a bootstrap test was used to evaluate if the AUC of ROC curve of the Flow Index was statistically different.

To validate the obtained results, a stratified tenfold cross-validation was used and best thresholds, sensitivity, specificity, PPV and NPV were reported as mean (SD) for the tenfold obtained.

A sensitivity analysis with Pressure Time Product (PTP) as a grouping variable, both with and without threshold load (called PTPtot and PTP, respectively), was also undertaken, using a value > 200 cmH2O s min−1 to identify high effort and a value of < 50cmH2O s min−1 to identify low effort [10].

The independent association between Flow Index and the respiratory system and lung mechanical properties (compliance and resistance) was evaluated using two linear mixed models, adjusting for Pmusc: Flow Index was the dependent variable, Pmusc and resistance or compliance were explanatory variables.

All reported tests were 2-sided and a P value of less than 0.05 was considered significant. Statistical analyses were performed with R (R Core Team, 2020. R Foundation for Statistical Computing, Vienna, Austria) with packages “pROC” (version 1.17.0.1) and “lme4” (version 1.1-26) [23,24,25].

Results

Twenty-eight patients were enrolled in the study. Four patients were excluded because the Pes trace was not reliable, therefore the data collected from 24 subjects was included in the analysis, accounting for a total of 702 breaths. Patients' characteristics on the day of enrollment and diagnosis at ICU admission are displayed in Table 1. Modification of PS resulted in a significant change in Pmusc, Flow Index, RR, RR/VT, VT/kg of IBW, P0.1, PTP and PTPtot, while minute ventilation did not change. As can be seen in Table 2, the levels of PS were different at high (Pmusc > 10 cmH2O), intermediate (Pmusc > 5 and < 10 cmH2O) and low effort (Pmusc < 5 cmH2O).

Table 1 Patients’ characteristics
Table 2 Ventilatory parameters at the three levels of PSV

It was possible to estimate respiratory system compliance and resistance in all but three patients [17], whereas lung compliance and resistance were calculated for all patients. Both lung and respiratory system mechanical properties were not independently associated with Flow Index (P values 0.91 and 0.36 for respiratory system compliance and resistance, and 0.20 and 0.09 for lung compliance and resistance, respectively).

Pmusc, P0.1, PTP, PTPtot and Flow Index differed between breaths with high, low and intermediate inspiratory effort, while RR, RR/VT and VT/kg of IBW did not differ in a statistically significant way (Table 3).

Table 3 Ventilatory parameters at high, intermediate and low inspiratory effort

Performance of Flow Index in detecting high inspiratory effort

Sixty-seven (10%) breaths were classified as breaths taken with high inspiratory effort. Respiratory variables during breaths taken at high inspiratory effort are displayed in Table 3. Flow Index was also significantly better at detecting high inspiratory effort compared to RR, RR/VT, VT/kg of IBW and P0.1 (P < 0.001 for all comparisons) as shown in Table 4 and Fig. 3A. The best threshold for the Flow Index was 4.5 (95% CI 3.0–5.1) with an AUC-ROC of 0.89 (95% CI 0.85–0.93) and an NPV of 0.98 (95% CI 0.97–0.99).

Table 4 Performance of the Flow Index and other routinely used parameters in etecting breaths with high inspiratory effort
Fig. 3
figure 3

Areas under the receiver operating curve for detecting high (A) and low (B) inspiratory effort. Left panel: Areas under the receiver operating characteristic curves for detecting high inspiratory effort, defined as Pmusc > 10 cmH2O. Flow Index is shown as a red curve, P0.1 as a yellow curve, respiratory rate as a pink curve, respiratory rate divided by tidal volume as a green curve, and tidal volume per kg of ideal body weight as a violet curve. Right panel: Receiver operating characteristic curve for detecting low inspiratory effort, defined as Pmusc < 5 cmH2O. Abbreviation AUC, area under the receiver operating characteristic curve; CI, confidence interval; P0.1, airway occlusion pressure; RR, respiratory rate; VT, tidal volume; IBW, ideal body weight

The internal tenfold cross validation yielded the same threshold for Flow Index, without significant modification in sensitivity, specificity, PPV and NPV (Additional file 1: Table S1).

Performance of Flow Index in detecting low inspiratory effort

Four hundred and forty breaths (63%) recorded during the study protocol were classified as breaths with low inspiratory effort.

The AUC-ROC for Flow Index was 0.80 (95% CI 0.76–0.83), significantly greater than the AUC for RR, RR/VT, VT/kg of IBW P0.1 (Fig. 3B). A Flow Index lower than 2.6 (95% CI 2.2–2.9) predicted low inspiratory effort with a probability of 0.84 (95% CI 0.81–0.87), while a Flow Index greater than this threshold excluded low inspiratory effort with a probability of 0.66 (95% CI 0.61–0.71) (Table 5).

Table 5 Performance of the FI and other parameters in detecting breaths with low inspiratory effort

The tenfold cross validation confirmed this finding, estimating the same best threshold. Sensitivity, specificity, PPV and NPV were not altered (Additional file 1: Table S2).

Results of secondary analyses with PTP as grouping variables confirm the results of the primary analysis, except for the fact that Flow index and P0.1 have a similar diagnostic accuracy. (see Additional file 1: Tables S3, S4, S5 and Figures S1, S2).

Results of secondary analyses with PTPtot as grouping variables show that P0.1 has a greater AUC-ROC than Flow Index. (see Additional file 1: Tables S6, S7, S8 and Figures S3, S4).

Discussion

This study shows that the Flow Index is particularly accurate in identifying low inspiratory effort during PSV and in excluding high inspiratory effort, while it is less precise in discriminating against the occurrence of high inspiratory effort.

The rationale behind the Flow Index arises from the knowledge that the inspiratory flow is driven by the difference of pressure between airway opening and the alveoli. In the presence of a constant inspiratory pressure and in absence of patient effort, the flow is maximal at the beginning of inspiration and decreases exponentially, due to the decreasing pressure gradient between the airway opening and the alveoli, adopting an upward concavity.

On the other hand, in the presence of a sustained patient inspiratory effort, the fall in pleural pressure due to the patient's muscle activity decreases alveolar pressure promoting the instantaneous inspiratory flow. As a consequence, the shape of the flow waveform the active patient takes a downward concavity. The inspiratory waveform profile is quantified by the Flow Index whose value is proportional to the activation of the inspiratory muscles. In fact the equation used to calculate the Flow Index mirrors the one used to calculate the Stress Index on the airway pressure waveform [18]. We recently showed that the Flow Index correctly identifies the shape of inspiratory flow and that the Flow index value is proportional to the patient's inspiratory effort. Mechanical properties of the lung and respiratory system were not significantly associated with Flow Index in our analysis, strengthening the assumption that the shape of the portion of inspiratory flow analyzed by Flow index is dependent mainly on patient-ventilator interaction.

The inspiratory effort can have two components: one preceding the beginning of the inspiratory flow, which is not modified by the applied inspiratory pressure, and one following the beginning of the inspiratory flow, which is modified by the level of applied inspiratory support. Of note, the Flow Index is influenced by definition only by the inspiratory effort performed after the inspiratory trigger activation, and does not take into account the eventual threshold load due to auto PEEP and trigger sensitivity. Moreover, since the portion of the inspiratory flow-time waveform analyzed by Flow index is included between the end of the ramp and before the expiratory trigger, Flow Index is not influenced by the cycling-off criterion.

Having previously shown that the Flow Index is correlated to patient inspiratory effort on a breath to breath basis, the primary aim of the present study was to establish Flow Index’s cutoff values in order to identify breaths with low (i.e. Pmusc lower than 5 cmH2O) or high (i.e. Pmusc greater than 10 cmH2O) spontaneous breathing effort and thus identify the “normal” Flow index range in clinical practice. We found that the Flow Index accurately identifies low inspiratory effort when it is lower than or equal to 2.5 and, on the other hand, allows to rule out high inspiratory effort when it is lower than 4.5. As noted by its relatively low positive predictive value, a high effort could be absent in the presence of a Flow Index greater than 4.5. This could be explained by the fact that from a theoretical point of view the Flow Index is affected by the distribution of the total respiratory work between patient and ventilator [14]. Therefore, the Flow Index may be higher than 4,5 despite a Pmusc of less than 10 cmH2O if the pressure applied by the ventilator is low, indicating that most of the work of breathing is performed by the patient. This notwithstanding, Flow Index still performed better in detecting high patient effort than all other analyzed indicators (i.e. RR, VT,RR/VT and P0.1).

Our data show that the performance of RR, VT and of RR/VT in identifying high or low inspiratory effort during PSV is relatively poor, compared with the Flow Index. This is not surprising since, despite their widespread use in clinical practice in order to titrate PSV [8, 26,27,28], the performance of breathing pattern parameters has been seldom demonstrated except in physiological studies [29]. To our knowledge, only one clinically relevant study recently demonstrated that a relative bradypnea (i.e. less than 17 breaths/min) may accurately detect over-assistance [30].

In the past 25 years, several alternative methods to assess inspiratory effort have been proposed, namely the pressure muscle index (PMI), the least square fitting method, the inspiratory occlusion method (ΔPocc) and the airway occlusion pressure (P0.1) [9, 10, 20, 31, 32]. However, the PMI and least square fitting method did not show adequate accuracy in detecting the patient's inspiratory effort [20, 31, 32]. The ΔPocc, recently proposed to detect excessive inspiratory effort [9], cannot be continuously monitored and provides scarce information on the occurrence of low patient effort. P0.1 is an established measure to assess respiratory drive and is deemed as a surrogate to estimate patient effort [10]. Differently from the Flow Index, P0.1 evaluates the global inspiratory effort, both before and after the start of the inspiratory flow. While P0.1 gives better information on the neuro-ventilatory drive, the Flow index may be more suitable to assess the spontaneous inspiratory effort which is impacted by the changes in inspiratory support, since it explores the post-inspiratory trigger part of the inspiratory effort. In our cohort, when examining the inspiratory effort after the beginning of inspiratory flow on the single breath, Flow Index is a more robust indicator than P0.1. When evaluating effort over one minute (using PTP without threshold load), this difference is set aside, as predictable (the increase in respiratory rate increased PTP). Nonetheless, when including threshold load (using PTPtot), P0.1 is better than Flow Index in detecting abnormal effort. This makes us believe that P0.1 is better than Flow index in evaluating the total entity of inspiratory effort (including threshold load and respiratory rate), while Flow Index is better than P0.1 in evaluating inspiratory effort after the beginning of flow. We speculate that Flow Index might be more appropriate to assess the adequacy of the inspiratory support level (which acts only after the start of flow), while P0.1 might be more useful for a global assessment of the whole inspiratory effort, including threshold load. This hypothesis is in accordance with our data and with the physiological basis behind Flow Index and P0.1, but further studies are needed to confirm it and to assess the relationship between the information provided by these two indices, which likely should be deemed as complementary rather than alternative.

Other more recent techniques used to assess patient effort are diaphragm ultrasound [33] and continuous monitoring of the electrical activity of the diaphragm [34]. Yet, these are better suited for intermittent patient assessments and require a dedicated and costly catheter, respectively [35]. Moreover, the current data is insufficient to evaluate the diagnostic accuracy of both these methods for detecting high and low patient effort.

Flow Index has the potential to become a useful tool to assess patient effort in everyday clinical practice. Using a relatively simple software update, it could be implemented as a continuous measure to be visualized on the mechanical ventilator screen, similarly to the Stress Index [18, 36]. The bedside continuous availability at virtually no additional cost makes it a suitable marker for both high and low-resource settings.

This study has several limitations. It is a single center study, and its external validity needs to be assessed with further research in order for the results to be generalizable. Also, since the definitions of high and low patient inspiratory effort during PSV vary across the literature, different results might have been found for different cutoffs. Moreover, the Flow Index software has not yet been implemented on mechanical ventilators, therefore continuous bedside monitoring is not possible to date. In addition, the analyzed breaths were 702, with 9.8 analyzed breaths, on average, for each patient at each of the three PS levels. Even though this number of breaths is appropriate for the analysis conducted in this study (using the single breath as a statistical unit), it might not be sufficient to assess a stable patient inspiratory effort during a longer period of PSV. Further analyses are warranted in order to evaluate the performance of each patient’s mean Flow Index over a longer period of time, its impact on clinically meaningful outcome parameters and to validate the Flow Index as a clinical tool.

Conclusion

Flow Index is accurate in detecting, continuously and non-invasively, high and low spontaneous inspiratory effort during PSV. These data support its potential application in clinical practice.

Availability of data and materials

The datasets during and/or analysed during the current study available from the corresponding author on reasonable request.

Abbreviations

P es :

Esophageal pressure

P musc :

Pressure generated by respiratory muscles

RR:

Respiratory rate

V T :

Tidal volume

P 0.1 :

Airway occlusion pressure

ΔP occ :

Swing in airway pressure generated by respiratory muscle effort recorded during a brief airway opening occlusion at end-expiration

PSV:

Pressure support ventilation

ICU:

Intensive Care Unit

PS:

Pressure support

P aw :

Airway opening pressure

Pcw :

Static recoil pressure of the chest wall

E cw :

Measured chest wall elastance

P plat,es :

End-inspiratory plateau esophageal pressure

P exp,es :

End-expiratory plateau esophageal pressure

AUC-ROC:

Areas under the receiver operating characteristic curves

IBW:

Ideal body weight

PPV:

Positive predictive value

NPV:

Negative predictive value

PTP:

Pressure time product from the beginning of the inspiratory flow

PTPtot :

Pressure time product from the beginning of the inspiratory effort

CI:

Confidence interval

SD:

Standard deviation

PEEP:

Positive end expiratory pressure

PMI:

Pressure muscle index

References

  1. Yoshida T, Uchiyama A, Matsuura N, Mashimo T, Fujino Y. Spontaneous breathing during lung-protective ventilation in an experimental acute lung injury model: high transpulmonary pressure associated with strong spontaneous breathing effort may worsen lung injury. Crit Care Med. 2012;40:1578–85.

    Article  Google Scholar 

  2. Goligher EC, Brochard LJ, Reid WD, Fan E, Saarela O, Slutsky AS, et al. Diaphragmatic myotrauma: a mediator of prolonged ventilation and poor patient outcomes in acute respiratory failure. Lancet Respir Med. 2019;7:90–8.

    Article  Google Scholar 

  3. Goligher EC, Jonkman AH, Dianti J, Vaporidi K, Beitler JR, Patel BK, Yoshida T, Jaber S, Dres M, Mauri T, Bellani G, Demoule A, Brochard L, Heunks L. Clinical strategies for implementing lung and diaphragm-protective ventilation: avoiding insufficient and excessive effort. Intensive Care Med. 2020;46:2314–26.

    Article  CAS  Google Scholar 

  4. Telias I, Spadaro S. Techniques to monitor respiratory drive and inspiratory effort. Curr Opin Crit Care. 2020;26:3–10.

    Article  Google Scholar 

  5. Yoshida T, Brochard L. Esophageal pressure monitoring: why, when and how? Curr Opin Crit Care. 2018;24:216–22.

    Article  Google Scholar 

  6. the PLeUral pressure working Group (PLUG—Acute Respiratory Failure section of the European Society of Intensive Care Medicine), Mauri T, Yoshida T, Bellani G, Goligher EC, Carteaux G, Rittayamai N, Mojoli F, Chiumello D, Piquilloud L, Grasso S, Jubran A, Laghi F, Magder S, Pesenti A, Loring S, Gattinoni L, Talmor D, Blanch L, Amato M, Chen L, Brochard L, Mancebo J. Esophageal and transpulmonary pressure in the clinical setting: meaning, usefulness and perspectives. Intensive Care Med. 2016;42:1360–73.

    Article  Google Scholar 

  7. Karthika M, Al Enezi F, Pillai L, Arabi Y. Rapid shallow breathing index. Ann Thorac Med. 2016;11:167.

    Article  Google Scholar 

  8. Banner MJ, Kirby RR, Kirton OC, DeHaven CB, Blanch PB. Breathing frequency and pattern are poor predictors of work of breathing in patients receiving pressure support ventilation. Chest. 1995;108:1338–44.

    Article  CAS  Google Scholar 

  9. Bertoni M, Telias I, Urner M, Long M, Del Sorbo L, Fan E, Sinderby C, Beck J, Liu L, Qiu H, Wong J, Slutsky AS, Ferguson ND, Brochard LJ, Goligher EC. A novel non-invasive method to detect excessively high respiratory effort and dynamic transpulmonary driving pressure during mechanical ventilation. Crit Care. 2019;23:346.

    Article  Google Scholar 

  10. Telias I, Junhasavasdikul D, Rittayamai N, Piquilloud L, Chen L, Ferguson ND, Goligher EC, Brochard L. Airway occlusion pressure as an estimate of respiratory drive and inspiratory effort during assisted ventilation. Am J Respir Crit Care Med. 2020;201:1086–98.

    Article  Google Scholar 

  11. Beloncle F, Piquilloud L, Olivier P-Y, Vuillermoz A, Yvin E, Mercat A, et al. Accuracy of P0.1 measurements performed by ICU ventilators: a bench study. Ann Intensive Care. 2019;9:104.

    Article  Google Scholar 

  12. Bellani G, Grassi A, Sosio S, Foti G. Plateau and driving pressure in the presence of spontaneous breathing. Intensive Care Med. 2019;45:97–8.

    Article  Google Scholar 

  13. Grasselli G, Brioni M, Zanella A. Monitoring respiratory mechanics during assisted ventilation. Curr Opin Crit Care. 2020;26:11–7.

    Article  Google Scholar 

  14. Albani F, Pisani L, Ciabatti G, Fusina F, Buizza B, Granato A, Lippolis V, Aniballi E, Murgolo F, Rosano A, Latronico N, Antonelli M, Grasso S, Natalini G. Flow Index: a novel, non-invasive, continuous, quantitative method to evaluate patient inspiratory effort during pressure support ventilation. Crit Care. 2021;25:196.

    Article  Google Scholar 

  15. Boles J-M, Bion J, Connors A, Herridge M, Marsh B, Melot C, Pearl R, Silverman H, Stanchina M, Vieillard-Baron A, Welte T. Weaning from mechanical ventilation. Eur Respir J. 2007;29:1033–56.

    Article  Google Scholar 

  16. Uhl RR, Lewis FJ. Digital computer calculation of human pulmonary mechanics using a least squares fit technique. Comput Biomed Res. 1974;7:489–95.

    Article  CAS  Google Scholar 

  17. Iotti GA, Braschi A, Brunner JX, Smits T, Olivei M, Palo A, et al. Respiratory mechanics by least squares fitting in mechanically ventilated patients: applications during paralysis and during pressure support ventilation. Intensive Care Med. 1995;21:406–13.

    Article  CAS  Google Scholar 

  18. Grasso S, Terragni P, Mascia L, Fanelli V, Quintel M, Herrmann P, Hedenstierna G, Slutsky AS, Ranieri VM. Airway pressure-time curve profile (stress index) detects tidal recruitment/hyperinflation in experimental acute lung injury. Crit Care Med. 2004;32:1018–27.

    Article  Google Scholar 

  19. Carteaux G, Mancebo J, Mercat A, Dellamonica J, Richard J-CM, Aguirre-Bermeo H, Kouatchet A, Beduneau G, Thille AW, Brochard L. Bedside adjustment of proportional assist ventilation to target a predefined range of respiratory effort. Crit Care Med. 2013;41:2125–32.

    Article  Google Scholar 

  20. Natalini G, Buizza B, Granato A, Aniballi E, Pisani L, Ciabatti G, Lippolis V, Rosano A, Latronico N, Grasso S, Antonelli M, Bernardini A. Non-invasive assessment of respiratory muscle activity during pressure support ventilation: accuracy of end-inspiration occlusion and least square fitting methods. J Clin Monit Comput. 2021;35:913–21.

    Article  Google Scholar 

  21. Obuchowski NA, McClish DK. Sample size determination for diagnostic accuracy studies involving binormal ROC curve indices. Stat Med. 1997;16:1529–42.

    Article  CAS  Google Scholar 

  22. Liu H, Li G, Cumberland WG, Wu T. Testing statistical significance of the Area under a Receiving Operating Characteristics curve for repeated measures design with bootstrapping. J Data Sci. 2021;3:257–78.

    Article  Google Scholar 

  23. R Core Team. R: a language and environment for statistical computing. Vienna: R Foundation for Statistical Computing; 2021. https://www.R-project.org/.

  24. Robin X, Turck N, Hainard A, Tiberti N, Lisacek F, Sanchez J-C, Müller M. pROC: an open-source package for R and S+ to analyze and compare ROC curves. BMC Bioinform. 2011;12:77.

    Article  Google Scholar 

  25. Bates D, Mächler M, Bolker B, Walker S. Fitting linear mixed-effects models using lme4. J Stat Softw. 2015;67:1–48.

    Article  Google Scholar 

  26. Akoumianaki E, Vaporidi K, Georgopoulos D. The injurious effects of elevated or nonelevated respiratory rate during mechanical ventilation. Am J Respir Crit Care Med. 2019;199:149–57.

    Article  CAS  Google Scholar 

  27. Carteaux G, Millán-Guilarte T, De Prost N, Razazi K, Abid S, Thille AW, Schortgen F, Brochard L, Brun-Buisson C, Mekontso DA. Failure of noninvasive ventilation for de novo acute hypoxemic respiratory failure: role of tidal volume. Crit Care Med. 2016;44:282–90.

    Article  Google Scholar 

  28. Lilitsis E, Stamatopoulou V, Andrianakis E, Petraki A, Antonogiannaki E-M, Georgopoulos D, Vaporidi K, Kondili E. Inspiratory effort and breathing pattern change in response to varying the assist level: a physiological study. Respir Physiol Neurobiol. 2020;280:103474.

    Article  CAS  Google Scholar 

  29. Berger KI, Sorkin IB, Norman RG, Rapoport DM, Goldring RM. Mechanism of relief of tachypnea during pressure support ventilation. Chest. 1996;109:1320–7.

    Article  CAS  Google Scholar 

  30. Pletsch-Assuncao R, Caleffi Pereira M, Ferreira JG, Cardenas LZ, de Albuquerque ALP, de Carvalho CRR, Caruso P. Accuracy of invasive and noninvasive parameters for diagnosing ventilatory overassistance during pressure support ventilation. Crit Care Med. 2018;46:411–7.

    Article  Google Scholar 

  31. Iotti GA, Braschi A, Brunner JX, Palo A, Olivet MC. Noninvasive evaluation of instantaneous total mechanical activity of the respiratory muscles during pressure support ventilation. Chest. 1995;108:208–15.

    Article  CAS  Google Scholar 

  32. Foti G, Cereda M, Banfi G, Pelosi P, Fumagalli R, Pesenti A. End-inspiratory airway occlusion: a method to assess the pressure developed by inspiratory muscles in patients with acute lung injury undergoing pressure support. Am J Respir Crit Care Med. 1997;156:1210–6.

    Article  CAS  Google Scholar 

  33. Umbrello M, Formenti P, Longhi D, Galimberti A, Piva I, Pezzi A, Mistraletti G, Marini JJ, Iapichino G. Diaphragm ultrasound as indicator of respiratory effort in critically ill patients undergoing assisted mechanical ventilation: a pilot clinical study. Crit Care. 2015;19:161.

    Article  Google Scholar 

  34. Bellani G, Mauri T, Coppadoro A, Grasselli G, Patroniti N, Spadaro S, Sala V, Foti G, Pesenti A. Estimation of patient’s inspiratory effort from the electrical activity of the diaphragm. Crit Care Med. 2013;41:1483–91.

    Article  Google Scholar 

  35. Bertoni M, Spadaro S, Goligher EC. Monitoring patient respiratory effort during mechanical ventilation: lung and diaphragm-protective ventilation. Crit Care. 2020;24:106.

    Article  Google Scholar 

  36. Sun X-M, Chen G-Q, Chen K, Wang Y-M, He X, Huang H-W, Luo X-Y, Wang C-M, Shi Z-H, Xu M, Chen L, Fan E, Zhou J-X. Stress index can be accurately and reliably assessed by visually inspecting ventilator waveforms. Respir Care. 2018;63:1094–101.

    Article  Google Scholar 

Download references

Acknowledgements

Not applicable.

Funding

No funding was received for this study.

Author information

Authors and Affiliations

Authors

Contributions

All authors contributed to the data acquisition. F.A., F.F.,G.C, L.P, V.L. and G.N. drafted the manuscript. G.N. designed and supervised the study. F.A. and G.N conducted the data analysis. All authors contributed substantially to editing, revising and finalizing the manuscript before submission. All authors read and approved the final manuscript.

Corresponding author

Correspondence to Federica Fusina.

Ethics declarations

Ethics approval and consent to participate

The protocol was approved by the local ethical committee (Comitato Etico della Provincia di Brescia) with protocol number NP4622.

Consent for publication

Not applicable.

Competing interests

The authors declare that they have no competing interests.

Additional information

Publisher's Note

Springer Nature remains neutral with regard to jurisdictional claims in published maps and institutional affiliations.

Supplementary Information

Additional file 1.

Supplementary material.

Rights and permissions

Open Access This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.

Reprints and permissions