- Editorial
- Open Access
- Published:
Which component of mechanical power is most important in causing VILI?
Critical Care volume 24, Article number: 39 (2020)
Background: energy, power, and VILI
Repeated applications of tidal energy inflict lung damage (VILI) when stress and strain exceed the limits of tissue tolerance. Inflation work and energy are the products of pressure and volume, which are loosely associated with stress and strain, respectively. Three major pressure components contribute to tidal inflation energy: flow resistive pressure, tidal elastic (driving) pressure, and the static elastic pressure baseline set by PEEP [1, 2].
Although total inflation power, i.e., the energy per cycle multiplied by ventilating frequency, has been causally linked to VILI risk [3], not all combinations of the three pressure components of energy and frequency that sum to the same total power value are equally hazardous. For injury to occur, a second requirement (apart from the raw power total) is a precondition; enough strain per cycle must be imposed to sufficiently distort lung parenchyma or disrupt vulnerable elements of the extracellular matrix that comprise the interdependent supporting framework for the alveolar network. The strain actually experienced at the tissue level depends on local force-amplifying influences such as geometrical stress focusing and rate of parenchymal expansion [4]. Duration of exposure (which determines the cumulative number of high stress cycles) is clearly of importance once the injuring strain threshold has been breached. Finally, whether or not a given pattern of ventilation injures also depends on lung capacity and innate vulnerability of lung tissues to the applied mechanical stress. Thus, a healthy lung can withstand power loads much greater than do the relatively fragile but functional alveoli that make up the “baby lung” of ARDS [5].
Sub-components and thresholds of tidal energy
Regarding the monitored variables of ventilation that relate to total inflation pressure, active debate has developed as to whether the element with most influence is end-inspiratory static (plateau) pressure or its difference from PEEP, known as the driving pressure (ΔP) [6,7,8]. PEEP bears a U-shaped relationship to VILI risk, with low levels tending to reduce atelectasis and distribute stress [9, 10]. On the other hand, high PEEP sets an elevated platform from which widespread tidal overstretching of the parenchyma as well as hemodynamic compromise become increasingly likely [10]. Finally, neither PEEP, plateau pressure, nor their difference (ΔP) reflect the dynamics and dissipated energy of the tidal cycle.
While energy clearly must be expended to produce damaging strain, uncertainty exists regarding which component of the applied energy per inflation cycle deserves closest attention. As implied by the earlier discussion, the “energy envelope” of tidal inflation encloses a total pressure-volume area defined by the combination of the flow-resistive, tidal elastic and PEEP-associated sectors that comprise it (Fig. 1a). Although flow-resistive energy largely dissipates in pushing gas past the endotracheal tube and through conducting airways, rapid movements of parenchymal tissues induce greater strains than slower ones—both during inflation and deflation [4, 11, 12].
a Left Panel: Elastic and flow resistive components of the total energy per tidal cycle during passive constant flow. b Right Panel: An arbitrarily assumed threshold for damaging pressure divides total tidal energy into subcomponent blocks below and above the boundary that separates potentially non-injurious (blocks G, D, and E) from injurious contributors (blocks F, A, B, and C), respectively. Simple algebraic estimates for each can be developed from clinical data using pre-specified parameters of mechanics or empirically observed measurements at the bedside. Example: total power = f × [A + B + C + D + E + F + G] = V’E × (RV’ + VT/2c + PEEP) or alternatively: (V’E/2) × (2 Ppeak − Pplat + PEEP). Total Dynamic Excessive Power = f × [A + B + C] = V’E × {1 + [(PEEP − PT) × c]/VT} × ½ [(VT/c) + PEEP + PT] or alternatively: V’E × [(Pplat − PT)/(Pplat − PEEP)] × [(Pplat + PT)/2)]. Abbreviations: R = Resistance; c = Compliance
The influence of PEEP-related inflation energy often has been discounted as well, using the reasoning that the pressure baseline is eventually recovered at the end of the deflation phase [8, 13]. Whatever the perceived wisdom of that argument, PEEP undeniably elevates volume and strain over the resting values of their fully relaxed states. In so doing, raising PEEP requires that each inflation volume increment is achieved at higher absolute pressure and with greater energy input. The driving pressure (tidal elastic) component of inflation energy parallels the amplitude of tidal volume and only quantifies incremental work due to raising elastic pressure above baseline PEEP.
It must be recognized that VILI may manifest in various ways: e.g., as inflammation, cellular disruption, or focal hyperinflation, with different tidal pressure components contributing unequally to each [9, 10]. For an individual lung, thresholds for stress, strain, energy, and power must exist, below which their applications pose little threat of whichever kind (Fig. 1a). Such thresholds are conditioned by the underlying integrity of the tissue elements, by the stage and extent of injury, and by the topographical position of the individual lung unit under consideration (local stress environment). In concept, such thresholds migrate to lower or higher levels as disease progresses or recedes.
While input of cumulative total energy and total power have been experimentally demonstrated to be linked to VILI [3, 14, 15], exactly which delivered component or subcomponent of power best correlates with damage has not been clearly identified. For example, it might be argued that only energy that relates to driving pressure or total elastic pressure (the product of volume and absolute alveolar pressure, ΔP + PEEP) would best track stretch and VILI risk. Alternatively, the total kinetic energy applied in excess of PEEP (the product of volume and the sum of ΔP plus flow resistive pressure) might hold equal interest regarding inflammatory changes. Furthermore, if we assume the existence of a pressure-strain threshold for damage and take it into consideration, perhaps only energy applied in excess of that level would be VILI-relevant. A breakdown of total tidal energy by threshold pressure gives rise to clinically measurable “blocks” of energy during passive inflation with constant flow (Fig. 1b). As the search to define the safety limits of power application proceeds, one or more of these energy sub-blocks hold potential to sharpen the precision of setting an upper target for lung-protecting power.
Estimating components of inflation energy
In theory, with resistance and compliance known and a threshold assumed, relatively simple algebraic formulae could estimate energy components (“blocks”) of research or clinical interest that are defined by pressure-volume areas during passive inflation with constant flow. These formulae use reasonable simplifying assumptions and a geometric approach to analysis (Fig. 1b). The theoretical impact of a proposed change in ventilation on the components of this energy/power complex could then be numerically estimated before they are actually made. Using similar principles when only already measured pressures, flow, volume, and frequency and are at hand, a comparable rationale and method produce condensed formulae that estimate these same energy components, so as to characterize the pattern already in use or to serially track the evolution and resolution of the ergo-trauma determinants of greatest interest (Fig. 1b).
Availability of data and materials
Not applicable.
Abbreviations
- P peak :
-
Peak pressure
- P res :
-
Resistive pressure
- P plat :
-
Plateau pressure
- P T :
-
Threshold pressure
- ΔP :
-
Driving pressure
- V T :
-
Tidal volume
- V E :
-
Minute ventilation
- C:
-
Compliance
- V:
-
Airflow
- R:
-
Airflow resistance
References
- 1.
Marini JJ, Rocco PRM, Gattinoni L. Static and dynamic contributors to VILI in clinical practice: pressure, energy, and power. Am J Respir Crit Care Med. 2019; https://doi.org/10.1164/rccm.201908-1545CI.
- 2.
Silva PL, Ball L, Rocco PRM, Pelosi P. Power to mechanical power to minimize ventilator-induced lung injury? Intensive Care Med Exp. 2019;7(Suppl 1):38.
- 3.
Santos RS, Maia LA, Oliveira MV, Santos CL, Moraes L, Pinto EF, et al. Biologic impact of mechanical power at high and low tidal volumes in experimental mild acute respiratory distress syndrome. Anesthesiology. 2018;128(6):1193–206.
- 4.
Faffe DS, Zin WA. Lung parenchymal mechanics in health and disease. Physiol Rev. 2009;89(3):759–75.
- 5.
Gattinoni L, Marini JJ, Pesenti A, Quintel M, Mancebo J, Brochard L. The baby lung became an adult. Intensive Care Med. 2016;42:663–73.
- 6.
Jain SV, Kollisch-Singule M, Satalin J, Searles Q, Dombert L, Abdel-Razek O, et al. The role of high airway pressure and dynamic strain on ventilator-induced lung injury in a heterogeneous acute lung injury model. Intensive Care Med Exp. 2017;5(1):25.
- 7.
Vasques F, Duscio E, Pasticci I, Romitti F, Vassalli F, Quintel M, et al. Is the mechanical power the final word on ventilator-induced lung injury?-we are not sure. Ann Transl Med. 2018;6(19):395.
- 8.
Huhle R, Serpa Neto A, Schultz MJ, Gama de Abreu M. Is mechanical power the final word on ventilator-induced lung injury?-No. Ann Transl Med. 2018;6(19):394.
- 9.
Samary CS, Santos RS, Santos CL, Felix NS, Bentes M, Barboza T, et al. Biological impact of transpulmonary driving pressure in experimental acute respiratory distress syndrome. Anesthesiology. 2015;123(2):423–33.
- 10.
Collino F, Rapetti F, Vasques F, Maiolo G, Tonetti T, Romitti F, et al. Positive end-expiratory pressure and mechanical power. Anesthesiology. 2019;130:119–30.
- 11.
Protti A, Maraffi T, Milesi M, Votta E, Santini A, Pugni P, et al. Role of strain rate in the pathogenesis of ventilator-induced lung edema. Crit Care Med. 2016;44(9):e838–45.
- 12.
Schmidt J, Wenzel C, Spassov S, Borgmann S, Lin Z, Wollborn J, et al. Flow-controlled ventilation attenuates lung injury in a porcine model of acute respiratory distress syndrome: a preclinical randomized controlled study. Crit Care Med. 2019; https://doi.org/10.1097/CCM.0000000000004209.
- 13.
Nieman GF, Satalin J, Andrews P, Habashi NM, Gatto LA. Lung stress, strain, and energy load: engineering concepts to understand the mechanism of ventilator-induced lung injury (VILI). Intensive Care Med Exp. 2016;4(1):16.
- 14.
Gattinoni L, Tonetti T, Cressoni M, Cadringher P, Herrmann P, Moerer O, et al. Ventilator-related causes of lung injury: the mechanical power. Intensive Care Med. 2016;42(10):1567–75.
- 15.
Moraes L, Silva PL, Thompson A, Santos CL, Santos RS, Fernandes MVS, et al. Impact of different tidal volume levels at low mechanical power on ventilator-induced lung injury in rats. Front Physiol. 2018;9:318.
Acknowledgements
Not applicable.
Funding
Brazilian Council for Scientific and Technological Development (CNPq), Rio de Janeiro State Research Foundation (FAPERJ).
Author information
Affiliations
Contributions
Both authors equally contributed all intellectual content of the manuscript. Both authors read and approved the final draft.
Corresponding author
Ethics declarations
Ethics approval and consent to participate
Not applicable.
Consent for publication
Not applicable.
Competing interests
The authors declare that they have no competing interests.
Additional information
Publisher’s Note
Springer Nature remains neutral with regard to jurisdictional claims in published maps and institutional affiliations.
Rights and permissions
Open Access This article is distributed under the terms of the Creative Commons Attribution 4.0 International License (http://creativecommons.org/licenses/by/4.0/), which permits unrestricted use, distribution, and reproduction in any medium, provided you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons license, and indicate if changes were made. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated.
About this article
Cite this article
Marini, J.J., Rocco, P.R.M. Which component of mechanical power is most important in causing VILI?. Crit Care 24, 39 (2020). https://doi.org/10.1186/s13054-020-2747-4
Received:
Accepted:
Published:
Keywords
- Ventilator-induced lung injury
- Inflation energy
- Ventilating power
- ARDS