- Editorial
- Open Access
- Published:
How I optimize power to avoid VILI
Critical Care volume 23, Article number: 326 (2019)
Rationale
Mechanical ventilation is an inherently dynamic process. Nonetheless, tidal volume and static tidal airway pressures (plateau, PEEP, and their difference, the driving pressure) have long served as the primary variables guiding prevention of ventilator-induced lung injury (VILI). Despite their prominence in current practice, such non-dynamic pressures cannot act alone to inflict damage; a pressure must be paired with a volume change, thereby expending energy. More specifically, any instigator of damage couples pressure applied directly to the lung, i.e., transpulmonary pressure (stress), to the associated change of lung volume (strain). Because damage depends not only on the frequency of such pairings but also on the rate of tidal stress/strain development across the epithelium and within individual extracellular fibrils that oppose lung expansion, rapid flows accentuate VILI hazard [1, 2].
Until recently, breathing frequency and duration of exposure have been underemphasized as contributors to VILI. As experimental data now indicate [3], however, more attention should be directed toward ventilating power, i.e., the product of energy delivered per tidal cycle and ventilating frequency, adjusted for the capacity of the lung to tolerate it. Non-mechanical factors, such as high inspired oxygen fraction, raised body temperature, and increased trans-alveolar gradients of pulmonary vascular pressure, increase vulnerability to VILI [4].
The total inflation energy of each tidal cycle is expended in overcoming flow resistive and elastic forces, both static and dynamic [5] (Fig. 1). Total energy input per minute, usually defined as “power,” is the product of minute ventilation (VE) and the sum of these same three tidal pressure components (flow resistive (FR = flow × resistance), DP, and PEEP). Primary emphasis is correctly placed on inflation, but expiration should not be ignored. Allowing sudden tidal deflations may accentuate VILI, perhaps due primarily to abrupt discharge of parenchymal energy [6].
Airway pressure profile during inflation with constant flow. Under these conditions, time is an accurate surrogate for inspired volume. Ppeak and Pplat are the peak dynamic and static (“plateau”) pressures. Total PEEP is comprised of the set PEEP value and auto-PEEP. Areas A, B, and C correspond to the flow resistive, tidal elastic, and PEEP-related tidal energy components, respectively. The sum of these three areas defines the total inflation energy of the tidal cycle: FR + DP/2 + PEEPtot. Note that while PEEP-related elastic energy (area C) elevates the inflation stress and strain platform for driving pressure, its temporarily stored elastic energy dissipates completely across the exhalation valve and circuitry during deflation
Although enticing, the power hypothesis needs further refinement in order to guide VILI avoidance in clinical practice. In the “baby lung” of ARDS, for example, ventilating energy and power concentrate within a “container” with less capacity and fewer air channels to accept the gas charge [7]. This spatial concentration amplifies both the magnitude and velocity of stretching forces of the tidal breath. Moreover, the mechanically heterogeneous environment magnifies stresses at the junctions of closed and open units [8, 9]. At the bedside, relative capacity for ventilation, which reflects severity, can be roughly approximated by the ratio of measured to predicted respiratory compliance (Cobs/Cpred) [5].
It seems unlikely that all combinations of frequency, tidal volume, and pressure (flow resistive pressure, driving pressure, and PEEP) that sum to the same power value are equally dangerous to a given lung. Hypothetically and experimentally, a tidal strain threshold for damage initiation, however indistinct, must first be crossed [10]. That threshold level—parenchymal susceptibility to VILI—is high for intact healthy lungs and relates inversely to severity of lung disease [10]. Of the three tidal pressure constituents, the driving pressure components of energy and power relate most directly to VILI [5, 11]. Rising PEEP, though not contributing to VILI in direct proportion to its major contribution suggested by the power equation, elevates the strains associated with driving and plateau pressures closer to that injury threshold, which is lower for some lung regions than for others and decreases as VILI progresses in a positive feedback cycle. Independently of whether mode is flow or pressure regulated, both the breathing frequency and the duration of repeated exposures contribute to VILI, provided that the injuring strain threshold has been exceeded.
Priorities in controlling damaging power
The strategy adopted during the earliest phase of support is crucial to curtailing cumulative exposure to high-stress cycles and improving eventual outcome. My first priority is to reduce vulnerability to VILI and the body’s needs for ventilation, oxygenation, and blood flow. For example, when spontaneous breathing is vigorous, deep sedation and/or paralytics minimize transpulmonary pressures, ventilatory demand, and the work of breathing. Estimating lung capacity serves to assess both disease severity and heterogeneity. Severely affected patients, judged either by criteria of refractory or worsening hypoxemia, or very low compliance, are proned [12]. Prone positioning usually recruits airspaces and helps even the distributions of transpulmonary pressures and tissue stresses [13]. Simultaneously, improved O2 exchanging efficiency typically allows reduction of FIO2. Fever and agitation are also suppressed to lessen ventilation demand. Successfully addressing the underlying cause of respiratory distress shortens the duration of ventilator support.
Once demands and co-contributors to vulnerability are minimized, ventilator settings assume top priority. In descending order of importance, those settings are as follows: (1) DP and PEEP, (2) VE, and (3) inspiration to expiration (I:E) ratio and flow profile. Whether during noninvasive or invasive ventilation, I attend first to both transpulmonary plateau pressure and transpulmonary driving pressure, and then to ventilating frequency and VE. I monitor the latter as a key component of power and helpful indicator of ventilating efficiency. Allowing hypercapnea (to PaCO2 ≤ ≈60 mmHg) promotes reductions of the frequency, VE, and ventilating power. I do not target a “fully open lung” but use stepwise recruiting maneuvers and “decremental” approach to find the least PEEP that achieves an effective balance between recruitment and overdistention. Assuming passive inflation and a normal chest wall, I regulate tidal volume so as to keep DP ≤ 15 cmH2O and plateau pressure ≤ 27 cmH2O. With tidal pressures and frequency set, I next modulate inspiratory flow, keeping the I:E ratio adjusted between 1:1.5 and 1:1. When the patient is passive, a controlled constant flow profile is preferred to the decelerating flow profile of pressure control [11, 14]. Because excessive power may be central to VILI risk, I assess and try to downregulate each of power’s determinants, as indicated. I do not set a specific upper limit for ventilating power itself, however, nor feel confident about using power criteria alone for initiating extracorporeal CO2 removal, as injury thresholds vary and numerical guidance from high-quality clinical studies is not currently available. (A deteriorating clinical trajectory is my primary criterion.) Finally, to minimize cumulative power exposure and VILI risk, I frequently re-evaluate my assumptions regarding needs and levels of ongoing respiratory supports.
Availability of data and materials
Not applicable
Abbreviations
- VILI:
-
Ventilator-induced lung injury
- PEEP:
-
Positive end-expiratory pressure
- PEEPtot:
-
Applied PEEP + auto-PEEP
- DP:
-
Driving pressure = plateau pressure minus total PEEP
- F:
-
Mean inspiratory flow rate
- R:
-
Airway plus tissue resistance to airflow
- I:E:
-
Inspiratory to expiratory
- VE :
-
Minute ventilation
- Cobs :
-
Observed respiratory compliance
- Cpred :
-
Predicted respiratory compliance
References
Protti A, Maraffi T, Milesi M, Votta E, Santini A, Pugni P, Andreis DT, Nicosia F, Zannin E, Gatti S, Vaira V, Ferrero S, Gattinoni L. Role of strain rate in the pathogenesis of ventilator-induced lung edema. Crit Care Med. 2016;44(9):e838–45. https://doi.org/10.1097/CCM. 0000000.
Garcia CS, Abreu SC, Soares RM, Prota LF, Figueira RC, Morales MM, Capelozzi VL, Zin WA, Rocco PR. Pulmonary morphofunctional effects of mechanical ventilation with high inspiratory air flow. Crit Care Med, 232. 2008;36(1):–9.
Cressoni M, Gotti M, Chiurazzi C, Massari D, Algieri I, Amini M, Cammaroto A, Brioni M, Montaruli C, Nikolla K, Guanziroli M, Dondossola D, Gatti S, Valerio V, Vergani GL, Pugni P, Cadringher P, Gagliano N, Gattinoni L. Mechanical power and development of ventilator-induced lung injury. Anesthesiology. 2016;124:1100–8.
Marini JJ, Hotchkiss JR, Broccard AF. Bench-to-bedside review: microvascular and airspace linkage in ventilator-induced lung injury. Crit Care. 2003;7(6):435–44.
Marini JJ, Jaber S. Dynamic predictors of VILI risk: beyond the driving pressure. Intensive Care Med. 2016;42:1597–600.
Goebel U, Haberstroh J, Foerster K, et al. Flow-controlled expiration: a novel ventilation mode to attenuate experimental porcine lung injury. Br J Anaesth. 2014;113(3):474–83. https://doi.org/10.1093/bja/aeu058.
Gattinoni L, Marini JJ, Pesenti A, Quintel M, Mancebo J, Brochard L. The “baby lung” became an adult. Intensive Care Med. 2016;42(5):663–73.
Mead J, Takishima T, Leith D. Stress distribution in lungs: a model of pulmonary elasticity. J Appl Physiol. 1970;28:596–608.
Cressoni M, Chiurazzi C, Gotti M, et al. Lung inhomogeneities and time course of ventilator-induced mechanical injuries. Anesthesiology. 2015;123:618–27.
Protti A, Cressoni M, Santini A, Langer T, Mietto C, Febres D, Chierichetti M, Coppola S, Conte G, Gatti S, Leopardi O, Masson S, Lombardi L, Lazzerini M, Rampoldi E, Cadringher P, Gattinoni L. Lung stress and strain during mechanical ventilation: any safe threshold? Am J Respir Crit Care Med. 2011;183:1354–62.
Silva PL, Ball L, Rocco PRM, Pelosi P. Power to mechanical power to reduce ventilator induced lung injury? Intensive Care Med Exp. 2019;7(Suppl 1):38:1–11. https://doi.org/10.1186/s40635-019-0243-4.
Scholten EL, Beitler JR, Prisk GK, Malhotra A. Treatment of ARDS with prone positioning. Chest. 2017;151(1):215–24. https://doi.org/10.1016/j.chest.2016.06.032.
Gattinoni L, Taccone P, Carlesso E, Marini JJ. Prone position in acute respiratory distress syndrome: rationale, indications, and limits. Am J Respir Crit Care Med. 2013;188(11):1286–93. https://doi.org/10.1164/rccm.201308-1532CI.
Maeda Y, Fujino Y, Uchiyama A, et al. Effects of peak inspiratory flow on development of ventilator-induced lung injury in rabbits. Anesthesiology. 2004;101(3):722–8.
Funding
None
Author information
Affiliations
Contributions
The author read and approved the final manuscript.
Corresponding author
Ethics declarations
Ethics approval and consent to participate
None
Consent for publication
I give consent as the sole author of this original opinionated Editorial.
Competing interests
The author declares that he has no competing interests.
Additional information
Publisher’s Note
Springer Nature remains neutral with regard to jurisdictional claims in published maps and institutional affiliations.
Rights and permissions
Open Access This article is distributed under the terms of the Creative Commons Attribution 4.0 International License (http://creativecommons.org/licenses/by/4.0/), which permits unrestricted use, distribution, and reproduction in any medium, provided you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons license, and indicate if changes were made. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated.
About this article
Cite this article
Marini, J.J. How I optimize power to avoid VILI. Crit Care 23, 326 (2019). https://doi.org/10.1186/s13054-019-2638-8
Received:
Accepted:
Published:
DOI: https://doi.org/10.1186/s13054-019-2638-8
Keywords
- Ventilator-induced lung injury
- VILI
- ARDS
- Energy
- Power
- Mechanical ventilation