Abstract
Background
Extracorporeal carbon dioxide removal (ECCO2R) uses an extracorporeal circuit to directly remove carbon dioxide from the blood either in lieu of mechanical ventilation or in combination with it. While the potential benefits of the technology are leading to increasing use, there are very real risks associated with it. Several studies demonstrated major bleeding and clotting complications, often associated with hemolysis and poorer outcomes in patients receiving ECCO2R. A better understanding of the risks originating specifically from the rotary blood pump component of the circuit is urgently needed.
Methods
High-resolution computational fluid dynamics was used to calculate the hemodynamics and hemocompatibility of three current rotary blood pumps for various pump flow rates.
Results
The hydraulic efficiency dramatically decreases to 5–10% if operating at blood flow rates below 1 L/min, the pump internal flow recirculation rate increases 6–12-fold in these flow ranges, and adverse effects are increased due to multiple exposures to high shear stress. The deleterious consequences include a steep increase in hemolysis and destruction of platelets.
Conclusions
The role of blood pumps in contributing to adverse effects at the lower blood flow rates used during ECCO2R is shown here to be significant. Current rotary blood pumps should be used with caution if operated at blood flow rates below 2 L/min, because of significant and high recirculation, shear stress, and hemolysis. There is a clear and urgent need to design dedicated blood pumps which are optimized for blood flow rates in the range of 0.5–1.5 L/min.
Background
Extracorporeal life support (ECLS), which is comprised of extracorporeal membrane oxygenation (ECMO) and extracorporeal carbon dioxide removal (ECCO2R) [1], is an emerging technology in the field of respiratory medicine used for various indications, including the acute respiratory distress syndrome (ARDS) and acute exacerbations of chronic obstructive pulmonary disease (COPD), or as a bridge to lung transplantation [2,3,4,5,6,7,8]. Recently, the EOLIA trial demonstrated a survival benefit for patients treated with ECMO compared to standard of care in severe ARDS [9, 10]. However, extracorporeal systems have substantial side effects, in particular, bleeding or clotting may occur in many patients. The concept of ECCO2R has been proposed as a safer alternative to ECMO due to the lower blood flow rates and smaller cannulae used. However, greater safety has not been established, and recent studies demonstrate increased bleeding complications in patients treated with ECCO2R [5, 11].
Historically, ECCO2R systems were developed from renal replacement therapy (RRT) and driven by roller pumps [12,13,14] or from high-flow extracorporeal membrane oxygenation (ECMO) devices driven by rotary pumps; most of them were centrifugal blood pumps in recent years. Few systems were designed specifically for ECCO2R [15,16,17]. In patients with moderate-to-severe ARDS, the SUPERNOVA pilot trial recently demonstrated the feasibility of reducing the intensity of mechanical ventilation by applying ECCO2R, using three different extracorporeal devices with blood flow rates ranging from 300 to 1000 mL/min [2]. However, although all three systems were characterized as “ECCO2R” [18], there were distinct differences with regard to the efficacy of CO2 removal. Systems derived from RRT devices are limited in blood flow rates (usually up to 500 mL/min), whereas those that are derived from high-flow ECMO devices are, in general, not limited by the blood flow rate, but more by cannula (or catheter) size and membrane lung surface area. In daily clinical practice, systems operating at blood flow rates up to 500 mL/min remove CO2 on the order of 80 mL/min. This can be nearly doubled by doubling the blood flow rate, thereby accounting for approximately 50% of the CO2 production of an adult resting intensive care unit (ICU) patient [19,20,21,22]. Furthermore, ECMO therapy for neonatal and pediatric patients uses comparable blood flow rates with current rotary blood pumps.
Whereas the efficacy and technical determinants of ECCO2R for adults, or low-flow ECMO for neonatal and pediatric patients, are reasonably well characterized, studies have raised the issue of the safety of the treatment [5, 23]. Although the blood flow rates used in ECCO2R are lower, and the cannulae are typically smaller than in high-flow ECMO, bleeding, clotting, and acquired van Willebrand syndrome are nonetheless common complications, influencing the outcome of clinical trials. Of note, hemolysis is one of the major complications, leading to worsening of clinical outcomes and is independently associated with mortality [24,25,26]. Studies by Braune et al. [5] and Karagiannidis et al. [11] (rotary pumps), as well as del Sorbo et al. [6] (roller pump), demonstrate significant bleeding complications in patients with acute exacerbation of COPD supported with ECCO2R. Similar observations were reported in neonatal and pediatric patients [25]. Whereas the complications induced by the oxygenator may be reduced by choosing the most appropriate membrane lung [21], special attention should be given to the blood pumps used at these low blood flow rates. Although blood flow rates may easily be reduced in high-flow ECMO with current rotary pumps, even down to less than 500 mL/min, the flow characteristics change considerably. Rotary blood pumps are developed for a very specific design point, but not for a broad spectrum of blood flow rates from 0 to 8 L/min. The respective components of the pump are dimensioned for this design point to allow for optimal flow guidance, as loss-free and efficient as possible, which may be lost at lower blood flow rates.
An understanding of the capabilities and complications of blood pumps at lower blood flow rates is essential for upcoming clinical trials of ECCO2R for patients with ARDS and acute exacerbation of COPD. We therefore sought to investigate the behavior of current ECMO and ECCO2R blood pumps with regard to hemocompatibility when operating at low blood flow rates. Since computational fluid dynamics (CFD) has been proven to accurately predict the behavior of blood pumps [27,28,29,30,31], this dedicated method was used to simulate the behavior of three currently used rotary blood pumps across a wide flow range.
Material and methods
Detailed geometries of the Xenios DP3 (Xenios AG, Heilbronn, Germany), Getinge Rotaflow (Getinge, Gothenburg, Sweden), and LivaNova Revolution (London, UK) pumps were derived from micro-CT scans and manual measurements using computer-aided design. The meshing of the pump’s internal blood volume was determined with tetrahedral elements and refined prism layers at the walls yielding up to 15.2 million mesh elements. Transient result averaging of the simulation results was performed over two impeller revolutions following five revolutions to ensure transient stability. The unsteady Reynolds-averaged Navier-Stokes (RANS) momentum and mass equations were iteratively solved using the commercial element-based finite volume method (ebFVM) solver CFX (ANSYS CFX, ANSYS, Inc., Canonsburg, PA, USA) and the sliding mesh approach. The blood was modeled with a shear-dependent viscosity [32] and a density of 1059 kg m−3. Convergence was monitored by the scalar variable residuals and stabilized predictions of the simulation parameters of this study. Detailed information is provided in the online data supplement. To briefly summarized the following.
Operation range and evaluation parameters
The low blood flow operation ranged between 0.5 and 4 L/min and a lower (150 mmHg) and upper (250 mmHg) pressure head target for typical CO2 removal applications. Identical pressure head at a given pump flow was achieved following speed adjustments for each pump (Additional file 3).
Hydraulic efficiency, secondary flows, and recirculation ratio
The hydraulic efficiency indicates the amount of loss with the conversion of the rotating impeller mechanical energy into hydraulic energy. It is the quotient of hydraulic pump output power to the impeller or shaft power, which can be numerically computed as the product of pump flow rate (Q) and pressure rise (∆P) and the product of impeller torque (T) and angular impeller speed (ω). Of note, although the hydraulic efficiency is a useful indicator for the amount of loss during pump operation, a high hydraulic efficiency does not simultaneously imply high hemocompatibility.
Secondary flows through the gaps between the rotating impeller and stationary housing are essential for adequate washout and to prevent the blood from clotting (Fig. 1a). However, excessive secondary or gap flow leakage can sacrifice the pump’s hydraulic efficiency.
a Main (pump flow) and secondary flows and flow paths (top and bottom gap flows) that add up to the impeller flow exemplified using the geometry details of the DP3. b Hydraulic efficiency curves of the three blood pumps under study for two constant impeller speeds to realize the pressure head target of 150 mmHg (lower speed in each case) and 250 mmHg
The ratio between all pump internal backflow (also referred to as secondary flow) and pump flow is defined as the recirculation ratio and specifies how often the blood is recirculated within the pump before reaching the pump outlet.
Hemolysis index and shear stress
The hemolysis index, HI (%), describes the percentage of damaged red blood cells with ΔfHb as the increase of plasma-free hemoglobin and Hb as the total amount of red blood cells. Current hemolysis estimation models typically relate hemolysis to the scalar shear stress and exposure time texp through a power-law relationship [33]:
The three-dimensional shear stress within the pump was derived from the velocity field obtained from the numerical simulations of the blood flow. It is commonly approximated by a scalar viscous shear stress τscalar following the equation:
Sij is the strain rate tensor, and μ is the dynamic viscosity of the blood.
The hemolysis index (Eq. 3) was numerically determined for each pump, pump flow, and pressure target employing empirical constants derived for use in rotary blood pumps [31] (C = 1.745 × 10−6, α = 1.963 and β = 0.0762) after conversion to the following equation [34, 35]:
Of note, numerical blood damage models are under continuous development and cannot fully substitute for experimental hemolysis testing. Nevertheless, numerical hemolysis results show a high correlation with experimental hemolysis results and are a reasonable substitute in the comparative pump analysis of this study.
Platelets of 32 non-septic patients, treated with ECCO2R (blood flow rates < 2 L/min) for acute exacerbation of COPD or for ARDS, were retrospectively analyzed in our institution from 2014 to 2018.
Results
Additional file 1 demonstrates the typical clinical scenario and side effects of ECCO2R. Platelets in 32 non-septic patients, treated with ECCO2R (blood flow rates < 2 L/min) for acute exacerbation of COPD or for ARDS, dropped by nearly half on average from 242 ± 101 (× 1000/μL) on day 0 to 127 ± 48 (× 1000/μL) on day 13 (Additional file 1A). Additional file 1B demonstrates the typical appearance of clotting within the pump, inducing severe hemolysis as a side effect of the treatment. Three frequently used rotary blood pumps (DP3, Rotaflow, and Revolution) were therefore experimentally evaluated by means of high-resolution CFD.
The hydraulic efficiency of the three blood pumps is demonstrated in Fig. 1. Of note, with decreasing pump flows, all systems present decreasing hydraulic efficiencies towards lower blood flow rates. At 0.5 L/min, the efficiency of the DP3 is only 7% against 150 mmHg of pressure head and 6.2% against 250 mmHg of pressure head; likewise, the hydraulic efficiency of Rotaflow (5.5; 4.7%) and Revolution (3.2; 2.7%) dramatically decreased, barely reaching 12% efficiency at 1 L/min. The DP3 system shows the best hydraulic efficiency at low flows, while the efficiency curves of the Rotaflow show a better trend towards flow rates above 4 L/min.
Higher rotational speeds create an offset towards lower hydraulic efficiency for all systems, meaning that the amount of loss increases.
In regard to the recirculation of the blood within the pump, Fig. 2a and b demonstrate the absolute flow rates in the secondary flow gaps in comparison with the impeller flow at 0.5 L/min and 250 mmHg pressure head, and the resulting recirculation ratios respectively. Of note, pumps with suspended rotors characteristically have multiple internal flow paths. The primary or main flow path is designed to generate the pump’s pressure head and fluid flow, while secondary flow paths are required to physically separate rotating impeller components from the stationary ones associated with the casing and to washout necessary gaps and mechanical bearings. Although the pumps effectively pump only 0.5 L/min (main flow), much higher internal backflows exist within the secondary flow paths (Figs. 1a and 2a and Additional file 2). The backflows must be pumped effectively through the impeller in addition to the actual pump flow (main flow), creating very high impeller flows. In Fig. 2b, the ratio between all internal backflow and pump flow is shown by the recirculation ratio (Eq. 2) over pump flow for the low- and high-pressure head target. This ratio becomes increasingly unfavorable for lower pump flows. At 0.5 L/min, it reaches a ratio of 6:1 for the DP3, 10:1 for the Rotaflow, and 12:1 for the Revolution. This means that the blood is likely recirculated between 6 and 12 times within the pumps before reaching the outlet. For higher pump flows (e.g., 4 L/min), this ratio becomes more balanced (0.8–1.2).
Shear stress of blood components is the major side effect generated by rotary blood pumps. Figure 3a depicts the shear stress histograms for all three pumps above 5 Pa. The Revolution (filling volume of 55 mL, largest of the compared pumps) shows consistently higher blood volume distributed over the entire shear stress interval range (Fig. 3a) with particularly more blood volume associated with non-physiological shear stresses above 100 Pa (Fig. 3b). The DP3 (filling volume 18.1 mL) shows more blood volume associated with shear stress regions compared to the Rotaflow (filling volume 28.8 mL). For all three pumps, the associated volume increases with pump speed, which consequently means a redistribution of the blood volume between 0 and 5 Pa to higher shear stress intervals.
a Shear stress histograms for the three pump systems for 0.5 L/min, low- and high-pressure head (150 and 250 mmHg). The blood volume of impeller and secondary gaps associated with a certain shear stress interval (x-axis) is plotted (DP3, 9.5 mL; Rotaflow, 18.2 mL; Revolution, 48 mL). The shear stress interval between 0 and 5 Pa contains most of the associated volume and was not shown for an improved view. Figure 4b details the associated volume above 100 Pa. c Volume rendering of shear stresses above 50 Pa illustrating potential hotspots within the pumps
Representative examples of shear stress profiles along blood streamlines, which result from pump flows of 0.5 and 4 L/min, are shown in Fig. 4. The mean residence times through the pump head were calculated based on 1000 streamlines to provide adequate representation of the complex flow characteristics. Figure 4a and b illustrate how the reduction of the pump flow not only increases the average residence time non-linearly within all pumps, but also causes multiple opportunities for exposure to high shear stresses from the increased internal recirculation (as detailed in Fig. 3), which increase the risk of blood trauma. Hellums [36] showed experimentally that the platelet activation threshold follows a consistent curve over a wide range of conditions on the shear stress-exposure time plane. A platelet activation threshold for blood pumps is conventionally taken as 50 Pa, which corresponds to an estimated particle transit time through the pump of 0.1 s [31]. Higher transit times, as shown in Fig. 4a, might thus condition an even lower activation threshold and thus more platelet activation potential.
All pump systems show an increase in the hemolysis index (single-pass blood damage) at lower pump flows (Fig. 5). The Revolution appears particularly susceptible to hemolysis compared with the DP3 and the Rotaflow, and the hemolysis index trend towards smaller pump flows is characterized by the largest slope reaching values of approximately 0.005% for 0.5 L/min against 250 mmHg. The curves of DP3 and Rotaflow also increase less steeply, but still significantly, towards smaller pump flows (~ 0.002% for 0.5 L/min against 250 mmHg). Although less blood is pumped through the pump at low blood flow rates, the concentration of damaged blood cells is greatly increased.
Discussion
For the first time, the present comparative study demonstrates systematically the potentially deleterious effects of currently used rotary blood pumps when operated at blood flow rates below 2 L/min, as is done in the clinical use of ECCO2R or neonatal and pediatric ECMO applications. By means of CFD, we could demonstrate that (a) the hydraulic efficiency dramatically decreases to 5–10% if operating at blood flow rates below 1 L/min, (b) the recirculation rate increases 6–12-fold in these flow ranges, and (c) adverse effects are increased due to multiple exposures to high shear stress. The deleterious consequences include a steep increase in hemolysis and destruction of platelets.
The use of ECCO2R is rapidly growing, and it remains a promising application of ECLS for ARDS or acute exacerbations of COPD, although there is currently no clear clinical indication for which there is high-quality evidence. Several studies are ongoing or planned for both applications. Although the rationale for the indications is clear, and the prevailing theory is that ECCO2R should be safer than ECMO in clinical practice, a concerning number of side effects have been reported in feasibility studies. As an example, major bleeding events occurred in more than 50% of patients in a trial aimed at avoiding invasive mechanical ventilation in patients with acute exacerbations of COPD [5], although this group of patients is not typically prone to bleeding when compared with patients who have severe sepsis. Bleeding may occur from loss of fibrinogen in the setting of its binding to the oxygenator, as well as circuit components, including the blood pumps, affecting the number and function of platelets, as shown in these experiments. Our current data on recirculation, high shear stress, and hemolysis are in line with the observed side effects and are at least in part responsible for this effect. This is of major importance, since, for instance, hemolysis is independently associated with mortality in some groups of patients [25].
From an engineering perspective, operating current blood pumps at low blood flow rates leads to low hydraulic efficiencies aggravating shear stress-induced blood trauma (Figs. 2, 3, and 4). The general efficiency slope of all systems suggests that the maximum efficiency point was designed for higher blood flow rates. Therefore, for all three blood pumps studied, the use of low blood flow rates for ECCO2R means this use is considerably removed from the design point of the pumps, meaning the optimal use that the pumps were designed for. The backflows (Fig. 3) must be pumped effectively through the impeller in addition to the actual pump flow, indicating that low pump flow does not also imply low impeller flow. The internal recirculation as presented in Fig. 2 causes multiple exposures to high shear stresses that are not physiologic, especially in the secondary gaps. All secondary flow paths induce fluid flow usually involving low volumetric flow rates and high shear stresses [37]. Given this, the ratio between the main flow and secondary flow at low flow rates might be causally related to the elevated complication risk. All pump systems show an increase of the hemolysis index when operated at blood flow rates below 2 L/min, which is further aggravated below 1 L/min. This is assumed to be a result of (a) the increased residence time of the blood within the pump, in the setting of reducing the pump flow itself and (b) unfavorable internal recirculation (Fig. 2), in combination with (c) multiple exposures to the respective shear stresses (Figs. 3 and 4) of the pump systems considered in this study. The results indicate a fundamental problem of hemocompatibility of all tested pumps for the low-flow operation as used for current ECCO2R applications.
Therefore, the concept of ECCO2R, which has been proposed as a safer alternative to ECMO due to the lower blood flow rates and smaller cannulae, used is questionable. In fact, the degree of adverse effects attributable to ECCO2R in clinical trials has been notably high, belying this notion. The role of blood pumps in contributing to adverse effects at the lower blood flow rates used during ECCO2R so far has not been well described. This study demonstrates that, at least in the case of the three pumps studied here, the role is significant. Current rotary blood pumps, such as the DP3, Rotaflow, or Revolution, should be used with caution if operated at blood flow rates below 2 L/min, because of significant and high recirculation, shear stress, and hemolysis.
Hemolysis, platelet function, and bleeding complications should be closely monitored in routine clinical practice and certainly within the context of clinical trials.
Limitations of the study
Blood damage models are under continuous development and subjected to certain limitations. The strength of current hemolysis models is the qualitative rather than the quantitative analysis. For example, in the context of a high blood recirculation, important correlations such as the cell damage history, which might influence the way a blood cell reacts when exposed to shear stress, are not taken into account. However, numerical predictions and experimentally determined hemolysis results show a very high correlation [38]. Moreover, this study focuses on three frequently used rotary blood pumps. Other rotary pumps or different pump systems (e.g., roller pumps) were not tested and may behave differently. Further experimental hemolysis testing of low pump flows is therefore advised to also illustrate quantitative differences in the hemolytic performance of the pumps considered in this study and other pump systems in general. However, our results are in line with recent data of flow-induced platelet activation, also demonstrating pump thrombogenicity due to long residence time [39].
Conclusions
The role of blood pumps in contributing to adverse effects at the lower blood flow rates used during ECCO2R is shown to be significant in this study. Current rotary blood pumps should be used with caution if operated at blood flow rates below 2 L/min, because of significant and high recirculation, shear stress, and hemolysis. There is a clear and urgent need to design dedicated blood pumps for ECCO2R and neonatal/pediatric ECMO applications, which are optimized for blood flow rates in the range of 0.5–1.5 L/min.
Availability of data and materials
All data generated or analyzed during this study are included in this published article.
Abbreviations
- CFD:
-
Computational fluid dynamics
- HI:
-
Hemolysis index
References
Brodie D, Slutsky AS, Combes A. Extracorporeal life support for adults with respiratory failure and related indications: a review. JAMA. 2019;322(6):557–68.
Combes A, Fanelli V, Pham T, Ranieri VM, European Society of Intensive Care Medicine Trials G, the “Strategy of Ultra-Protective lung ventilation with Extracorporeal CORfN-OmtsAi”. Feasibility and safety of extracorporeal CO2 removal to enhance protective ventilation in acute respiratory distress syndrome: the SUPERNOVA study. Intensive Care Med. 2019;45(5):592–600.
Brochard L, Slutsky A, Pesenti A. Mechanical ventilation to minimize progression of lung injury in acute respiratory failure. Am J Respir Crit Care Med. 2017;195(4):438–42.
Karagiannidis C, Brodie D, Strassmann S, Stoelben E, Philipp A, Bein T, Muller T, Windisch W. Extracorporeal membrane oxygenation: evolving epidemiology and mortality. Intensive Care Med. 2016;42(5):889–96.
Braune S, Sieweke A, Brettner F, Staudinger T, Joannidis M, Verbrugge S, Frings D, Nierhaus A, Wegscheider K, Kluge S. The feasibility and safety of extracorporeal carbon dioxide removal to avoid intubation in patients with COPD unresponsive to noninvasive ventilation for acute hypercapnic respiratory failure (ECLAIR study): multicentre case-control study. Intensive Care Med. 2016;42(9):1437–44.
Del Sorbo L, Pisani L, Filippini C, Fanelli V, Fasano L, Terragni P, Dell’Amore A, Urbino R, Mascia L, Evangelista A, et al. Extracorporeal Co2 removal in hypercapnic patients at risk of noninvasive ventilation failure: a matched cohort study with historical control. Crit Care Med. 2015;43(1):120–7.
Boyle AJ, Sklar MC, McNamee JJ, Brodie D, Slutsky AS, Brochard L, McAuley DF, International EN. Extracorporeal carbon dioxide removal for lowering the risk of mechanical ventilation: research questions and clinical potential for the future. Lancet Respir Med. 2018;6(11):874–84.
Biscotti M, Gannon WD, Agerstrand C, Abrams D, Sonett J, Brodie D, Bacchetta M. Awake extracorporeal membrane oxygenation as bridge to lung transplantation: a 9-year experience. Ann Thorac Surg. 2017;104(2):412–9.
Combes A, Hajage D, Capellier G, Demoule A, Lavoue S, Guervilly C, Da Silva D, Zafrani L, Tirot P, Veber B, et al. Extracorporeal membrane oxygenation for severe acute respiratory distress syndrome. N Engl J Med. 2018;378(21):1965–75.
Goligher EC, Tomlinson G, Hajage D, Wijeysundera DN, Fan E, Juni P, Brodie D, Slutsky AS, Combes A. Extracorporeal membrane oxygenation for severe acute respiratory distress syndrome and posterior probability of mortality benefit in a post hoc Bayesian analysis of a randomized clinical trial. JAMA. 2018;320(21):2251–9.
Karagiannidis C, Strassmann S, Schwarz S, Merten M, Fan E, Beck J, Sinderby C, Windisch W. Control of respiratory drive by extracorporeal CO2 removal in acute exacerbation of COPD breathing on non-invasive NAVA. Crit Care. 2019;23(1):135.
Allardet-Servent J, Castanier M, Signouret T, Soundaravelou R, Lepidi A, Seghboyan JM. Safety and efficacy of combined extracorporeal CO2 removal and renal replacement therapy in patients with acute respiratory distress syndrome and acute kidney injury: the pulmonary and renal support in acute respiratory distress syndrome study. Crit Care Med. 2015;43(12):2570–81.
Schmidt M, Jaber S, Zogheib E, Godet T, Capellier G, Combes A. Feasibility and safety of low-flow extracorporeal CO2 removal managed with a renal replacement platform to enhance lung-protective ventilation of patients with mild-to-moderate ARDS. Crit Care. 2018;22(1):122.
Nentwich J, Wichmann D, Kluge S, Lindau S, Mutlak H, John S. Low-flow CO2 removal in combination with renal replacement therapy effectively reduces ventilation requirements in hypercapnic patients: a pilot study. Ann Intensive Care. 2019;9(1):3.
Zanella A, Castagna L, Salerno D, Scaravilli V, Abd El Aziz El Sayed Deab S, Magni F, Giani M, Mazzola S, Albertini M, Patroniti N, et al. Respiratory electrodialysis. A novel, highly efficient extracorporeal CO2 removal technique. Am J Respir Crit Care Med. 2015;192(6):719–26.
Burki NK, Mani RK, Herth FJ, Schmidt W, Teschler H, Bonin F, Becker H, Randerath WJ, Stieglitz S, Hagmeyer L, et al. A novel extracorporeal CO (2) removal system: results of a pilot study of hypercapnic respiratory failure in patients with COPD. Chest. 2013;143(3):678–86.
Batchinsky AI, Jordan BS, Regn D, Necsoiu C, Federspiel WJ, Morris MJ, Cancio LC. Respiratory dialysis: reduction in dependence on mechanical ventilation by venovenous extracorporeal CO2 removal. Crit Care Med. 2011;39(6):1382–7.
Conrad SA, Broman LM, Taccone FS, Lorusso R, Malfertheiner MV, Pappalardo F, Nardo MD, Belliato M, Grazioli L, Barbaro RP, et al. The Extracorporeal Life Support Organization Maastricht Treaty for Nomenclature in Extracorporeal Life Support. A position paper of the extracorporeal life support organization. Am J Respir Crit Care Med. 2018;198(4):447–51.
Karagiannidis C, Hesselmann F, Fan E. Physiological and technical considerations of extracorporeal CO2 removal. Crit Care. 2019;23(1):75.
Karagiannidis C, Kampe KA, Sipmann FS, Larsson A, Hedenstierna G, Windisch W, Mueller T. Veno-venous extracorporeal CO2 removal for the treatment of severe respiratory acidosis: pathophysiological and technical considerations. Crit Care. 2014;18(3):R124.
Karagiannidis C, Strassmann S, Brodie D, Ritter P, Larsson A, Borchardt R, Windisch W. Impact of membrane lung surface area and blood flow on extracorporeal CO2 removal during severe respiratory acidosis. Intensive Care Med Exp. 2017;5(1):34.
Strassmann S, Merten M, Schafer S, de Moll J, Brodie D, Larsson A, Windisch W, Karagiannidis C. Impact of sweep gas flow on extracorporeal CO2 removal (ECCO2R). Intensive Care Med Exp. 2019;7(1):17.
Sklar MC, Beloncle F, Katsios CM, Brochard L, Friedrich JO. Extracorporeal carbon dioxide removal in patients with chronic obstructive pulmonary disease: a systematic review. Intensive Care Med. 2015;41(10):1752–62.
Omar HR, Mirsaeidi M, Socias S, Sprenker C, Caldeira C, Camporesi EM, Mangar D. Plasma free hemoglobin is an independent predictor of mortality among patients on extracorporeal membrane oxygenation support. PLoS One. 2015;10(4):e0124034.
Okochi S, Cheung EW, Barton S, Zenilman A, Shakoor A, Street C, Streltsova S, Chan C, Brewer MP, Middlesworth W. An analysis of risk factors for hemolysis in children on extracorporeal membrane oxygenation. Pediatr Crit Care Med. 2018;19(11):1059–66.
Lehle K, Philipp A, Zeman F, Lunz D, Lubnow M, Wendel HP, Gobolos L, Schmid C, Muller T. Technical-induced hemolysis in patients with respiratory failure supported with Veno-venous ECMO - prevalence and risk factors. PLoS One. 2015;10(11):e0143527.
Fraser KH, Taskin ME, Griffith BP, Wu ZJ. The use of computational fluid dynamics in the development of ventricular assist devices. Med Eng Phys. 2011;33(3):263–80.
Taskin ME, Fraser KH, Zhang T, Gellman B, Fleischli A, Dasse KA, Griffith BP, Wu ZJ. Computational characterization of flow and hemolytic performance of the UltraMag blood pump for circulatory support. Artif Organs. 2010;34(12):1099–113.
Zhang J, Zhang P, Fraser KH, Griffith BP, Wu ZJ. Comparison and experimental validation of fluid dynamic numerical models for a clinical ventricular assist device. Artif Organs. 2013;37(4):380–9.
Burgreen GW, Antaki JF, Wu ZJ, Holmes AJ. Computational fluid dynamics as a development tool for rotary blood pumps. Artif Organs. 2001;25(5):336–40.
Fraser KH, Zhang T, Taskin ME, Griffith BP, Wu ZJ. A quantitative comparison of mechanical blood damage parameters in rotary ventricular assist devices: shear stress, exposure time and hemolysis index. J Biomech Eng. 2012;134(8):081002.
Ballyk PD, Steinman DA, Ethier CR. Simulation of non-Newtonian blood flow in an end-to-side anastomosis. Biorheology. 1994;31(5):565–86.
Giersiepen M, Wurzinger LJ, Opitz R, Reul H. Estimation of shear stress-related blood damage in heart valve prostheses--in vitro comparison of 25 aortic valves. Int J Artif Organs. 1990;13(5):300–6.
Garon A, Farinas MI. Fast three-dimensional numerical hemolysis approximation. Artif Organs. 2004;28(11):1016–25.
Farinas MI, Garon A, Lacasse D, N’Dri D. Asymptotically consistent numerical approximation of hemolysis. J Biomech Eng. 2006;128(5):688–96.
Hellums JD. 1993 Whitaker Lecture: biorheology in thrombosis research. Ann Biomed Eng. 1994;22(5):445–55.
Zhang J, Gellman B, Koert A, Dasse KA, Gilbert RJ, Griffith BP, Wu ZJ. Computational and experimental evaluation of the fluid dynamics and hemocompatibility of the CentriMag blood pump. Artif Organs. 2006;30(3):168–77.
Taskin ME, Fraser KH, Zhang T, Wu C, Griffith BP, Wu ZJ. Evaluation of Eulerian and Lagrangian models for hemolysis estimation. ASAIO J. 2012;58(4):363–72.
Fuchs G, Berg N, Broman LM, Prahl Wittberg L. Flow-induced platelet activation in components of the extracorporeal membrane oxygenation circuit. Sci Rep. 2018;8(1):13985.
Acknowledgements
We are grateful to Marek Weiler, Institute for Experimental Molecular Imaging, Medical Faculty, RWTH Aachen University, Germany, for helping with the micro-CT scans.
Funding
This study was in part funded by the German Federal Ministry of Education and Research No. 13GW0219B.
Author information
Authors and Affiliations
Contributions
SG, FH, and CK designed the study. SG and FH performed the CFD calculation. SG wrote the main draft of the manuscript. All authors contributed to the final drafting of the manuscript and read and approved the final manuscript.
Corresponding author
Ethics declarations
Ethics approval and consent to participate
Not applicable
Consent for publication
The manuscript has been read and its submission approved by all co-authors.
Competing interests
CK received travel grants and lecture fees from Maquet, Rastatt, Germany. WW received fees for advisory board meetings and lectures from Maquet Cardiopulmonary, Rastatt, Germany. CK and WW received an open research grant for the hospital from Maquet Cardiopulmonary, Rastatt, Germany. DB reports serving as the co-chair of the trial steering committee for the VENT-AVOID trial sponsored by ALung Technologies; serving on the medical advisory boards for Baxter, BREETHE, and Hemovent (unpaid); and previously serving on the medical advisory board of ALung Technologies. SGH is half-time employed at enmodes GmbH Aachen. FH, JA, LV, and US declare that they have no competing interests.
Additional information
Publisher’s Note
Springer Nature remains neutral with regard to jurisdictional claims in published maps and institutional affiliations.
Supplementary information
Additional file 1.
A: Platelet count trend over 13 days including n = 32 non-septic patients. B: Thrombosis formation indicated by arrows in the middle of the pump head.
Additional file 2.
Geometric representations of the DP3 (a), Rotaflow (b), and Revolution (c). Details of the mesh are provided for the DP3 and Revolution as insets (I + II) for a and b detailing the mesh of the respective gaps between impeller and casing.
Additional file 3.
Online Data Supplement.
Rights and permissions
Open Access This article is distributed under the terms of the Creative Commons Attribution 4.0 International License (http://creativecommons.org/licenses/by/4.0/), which permits unrestricted use, distribution, and reproduction in any medium, provided you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons license, and indicate if changes were made. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated.
About this article
Cite this article
Gross-Hardt, S., Hesselmann, F., Arens, J. et al. Low-flow assessment of current ECMO/ECCO2R rotary blood pumps and the potential effect on hemocompatibility. Crit Care 23, 348 (2019). https://doi.org/10.1186/s13054-019-2622-3
Received:
Accepted:
Published:
DOI: https://doi.org/10.1186/s13054-019-2622-3
Keywords
- ARDS
- ECMO
- ECCO2R
- ECLS
- Centrifugal blood pumps