Skip to main content

Automated continuous noninvasive ward monitoring: future directions and challenges

Abstract

Automated continuous noninvasive ward monitoring may enable subtle changes in vital signs to be recognized. There is already some evidence that automated ward monitoring can improve patient outcome. Before automated continuous noninvasive ward monitoring can be implemented in clinical routine, several challenges and problems need to be considered and resolved; these include the meticulous validation of the monitoring systems with regard to their measurement performance, minimization of artifacts and false alarms, integration and combined analysis of massive amounts of data including various vital signs, and technical problems regarding the connectivity of the systems.

Patient monitoring by definition is the repeated or continuous observation of vital signs or physiologic functions to ensure patient safety and guide therapeutic interventions. Today, most advanced cardiorespiratory monitoring systems depend on invasive sensors, cables, and bulky monitors to recognize, transfer, process, and display the bio-signals to be monitored. Therefore, continuous advanced cardiorespiratory monitoring is mainly restricted to the intensive care unit, the operating room, and the post anesthesia care unit. Most other monitoring in the hospital continues to be basic and intermittent—including monitoring on medical and surgical general care wards. When at home, before-and-after hospital admission, patients are usually not monitored at all [1].

While most advanced monitoring is in place in intensive care units, nearly half of all adverse events in hospitalized patients occur on the general care ward [2,3,4]. Ironically, this area—also referred to as “the patient’s room”—is traditionally regarded as a place of recovery for the more stable medical or surgical patients, who will (in the absence of setbacks) transition to leave the hospital. In addition, the European Surgical Outcomes Study (EuSOS) [5] revealed that about three quarters of patients who died in the hospital after surgery were not admitted to an intensive care unit at any stage after surgery; this indicates that the general care ward plays a pivotal role in the care for patients in the postoperative period, a period in which patients are especially prone to developing clinical deterioration and life-threatening complications [5, 6]. Not only are catastrophic cardiorespiratory events common in general care ward environments, their outcomes are significantly worse compared with similar events in monitored intensive care units. For example, a large national registry identified 44,551 index events across more than 300 US hospitals [7]. More importantly these acute respiratory events on inpatient wards had an associated in-hospital mortality of approximately 40% [7].

Current ward monitoring protocols typically consist of intermittent spot checks by a nurse about every 4–8 h. This leaves patients unmonitored for most of the time during their hospital stay [8]. Alterations in vital signs as warning signs of clinical deterioration are frequently not or only belatedly recognized by the conventional spot check-based monitoring strategy. In hospitalized patients recovering from non-cardiac surgery, severe prolonged hypoxemia is common and unfortunately seriously underestimated using intermittent vital sign checks [9, 10]. Similar patterns regarding the rate of recognized abnormal vital signs were observed for tachycardia, bradycardia, tachypnea, and bradypnea [10]. In patients recovering from abdominal surgery on the general care ward, postoperative hypotension (mean arterial pressure < 65 mmHg for ≥ 15 min) has been shown to occur in about one fifth of patients and not to be recognized by routine vital sign assessments in about half of the cases [11]. In addition to missing critical changes in vital signs, the recognition of abnormal vital signs by a bedside nurse often triggers a long chain of commands resulting in delays until an intervention can be taken [12].

A closed claims analysis of opioid induced respiratory compromise on the general care ward identified nearly half of all these events occurred within 2 h of the last nursing check [13]. In addition, the authors concluded that nearly all of these events would have been prevented by better continuous monitoring and education [13]. Considering that most nursing spot checks of vital signs leave gaps of about 4 h in-between two consecutive assessments, this period is associated with the highest vulnerability.

Automated continuous noninvasive ward monitoring is a promising approach to closely follow changes in vital signs over time and thus identify patients who are deteriorating in a timely fashion (Fig. 1). The rationale behind continuous ward monitoring is that most hospitalized patients do not deteriorate all of a sudden. Although complications often become clinically apparent as acute cardiocirculatory or respiratory failure and acute changes in consciousness, we know for a long time that subtle abnormalities in vital signs usually precede these life-threatening conditions, sometimes by 6–12 h [12, 14,15,16]. Subtle changes in blood pressure, heart rate, respiratory rate, or oxygen saturation are early signs of clinical deterioration eventually leading to adverse events [12, 16]. Automated continuous noninvasive ward monitoring may enable a patient’s clinical deterioration to be identified well before a serious adverse event occurs [16]. Further, novel monitoring technologies also may enable advanced hemodynamic variables such as stroke volume, cardiac output, and dynamic cardiac preload parameters to be monitored continuously in patients on the general care ward [17,18,19]; to date, these variables—in contrast to surgical or critically ill patients—play no role in the treatment of patients in this environment.

Fig. 1
figure 1

Automated continuous noninvasive ward monitoring allows the healthcare provider to closely follow changes in vital signs over time and identify patients who are deteriorating earlier than conventional intermittent spot check monitoring. Early recognition of clinical deterioration enables rapid therapeutic interventions which may be life saving in certain situations

There is already some evidence that intensified and automated ward monitoring of vital signs may improve patient outcome by a reduction of rescue events [1]. Before-and-after studies showed that the deployment of an electronic automated advisory vital signs monitoring and notification system is associated with significant improvements in key patient-centered clinical outcomes in patients treated on the normal ward [20, 21]. In an orthopedic ward, the implementation of a continuous pulse oximetry surveillance system linked to a nursing notification system reduced the number of rescue events from 3.4 to 1.2 per 1000 patient discharges and also reduced the rate of intensive care unit transfers (before-and-after study) [22]. In another study, the implementation of a system allowing for continuous monitoring of heart and respiratory rate in a medical-surgical unit was associated with lower “code blue” rates (6.3 before to 0.9 after implementation per 1000 patients) [23].

Although continuous ward monitoring is not standard of care today, innovative monitoring systems—in theory—would already allow us to noninvasively and continuously monitor heart rate, blood pressure, respiratory rate, oxygen saturation, skin temperature, body posture, activity, and location within the hospital [24,25,26,27,28]. Battery powered, wearable or adhesive, wireless monitoring systems that communicate with mobile devices or patient monitors may—in the near future—give hospitalized patients the freedom to move within their rooms and the hospital while being monitored [25, 28]. While, the best-case scenario would be the implementation of universal continuous smart monitoring for all inpatients, there may also be value in attempting to identify the highest risk strata of those most likely to face sudden unprecedented episodes of cardiorespiratory compromise. Novel scores such as PRODIGY developed using continuous capnography and oximetry may help the perioperative clinician in early interventions using a combination of better monitoring and other proactive strategies to avert future problems [29].

Furthermore, before automated continuous noninvasive ward monitoring becomes a reality in routine clinical care outside of studies, several problems and limitations need to be considered. Most importantly, monitoring systems need to be reliable, accurate, and be able to provide readings of vital signs with a low rate of artifacts and false alarms. However, some of the currently available monitoring systems lack clinically acceptable accuracy and precision [30]. Therefore, meticulous validation needs to precede the use of novel cardiorespiratory monitoring systems in studies or clinical practice. Especially for blood pressure, a key hemodynamic variable, reliable continuous noninvasive monitoring is technically challenging and unavailable in most smart portable systems [27]. Some monitoring systems still suffer from high rates of artifactual readings and false alarms [31]. Not only is the problem of false alarms a nuisance, it will almost always lead to an increasing level of alarm fatigue within bedside providers, the so called first-responders to needless alarms. In this regard, some vital signs are more prone to artifacts and false alarms than others—for example, capnography as a measure of ventilation being one that has always been a prime suspect for this. Frequent and false alarms may be automatically identified and reduced by cross-checking and machine learning algorithms [32,33,34].

Automated continuous noninvasive ward monitoring of a variety of vital signs with one or more sensors will create a massive amount of data that need to be processed in real-time, stored, and secured. These data reflecting different bio-signals will need to be integrated and analyzed together to allow the identification of certain patterns of vital sign alterations instead of merely recognizing that single values of single variables are outside of their normal range. Several predictive statistical models have already been developed, validated, and embedded in electronic medical records as automated aggregated “early warning scores” that assign weights to altered vital signs proportionate to their deviation from normal ranges [35,36,37]. Identifying changes in physiologic variables over time and using machine and deep learning methods may improve the predictive capabilities of these risk stratification tools [38, 39].

Other challenges concern the technical connectivity between sensors and monitoring systems. While the “internet of things” (i.e., a network of devices, vehicles, and home appliances) became part of our daily life, wireless data transmission and processing are not yet well established in hospitals and other health care facilities. Problems for the implementation of wireless monitoring systems include—but are not limited to—range, power consumption, integration in electronic health records, and cybersecurity [1]. Legal issues regarding data protection and privacy rights are beyond the scope of this article but are essential topics that need to be taken care of before ward monitoring can be implemented in healthcare systems.

Finally, before automated continuous noninvasive ward monitoring can be recommended for routine clinical use, we need to await the results of adequately powered randomized controlled trials demonstrating its effectiveness in improving the quality of care and carefully chosen patient-centered outcomes. As a next step, research may then focus on investigating which patients benefit from expanding ward monitoring to home monitoring in the period after hospital discharge [1, 40].

Conclusions

Automated continuous noninvasive ward monitoring seems to be an intriguing opportunity to timely detect clinical problems by recognizing subtle changes in vital signs and improve patient outcomes on the general care ward. There is already some evidence—mainly from before-and-after studies—that automated ward monitoring can improve patient outcome. From a technical point of view, monitoring systems for automated continuous noninvasive ward monitoring are already available and will be further refined during the next years, probably resulting in small, wireless, and wearable sensors. However, before automated continuous noninvasive ward monitoring can be implemented in clinical routine, several challenges and problems need to be considered and resolved; these include the meticulous validation of the monitoring systems with regard to their measurement performance, minimization of artifacts and false alarms, integration and combined analysis of massive amounts of data including various vital signs, and technical problems regarding the connectivity of the systems. The primary scientific aim, though fairly simple, needs some thought and well-planned trial design, and would look to evaluate in robust and adequately powered trials whether automated continuous noninvasive ward monitoring can improve patient outcome compared with current standard spot-check monitoring. Till such time, it seems rather inappropriate to leave our patients under-monitored and unprotected for large periods of time as they recover from illness on our general care hospital wards.

Availability of data and materials

Not applicable

References

  1. McGillion MH, Duceppe E, Allan K, Marcucci M, Yang S, Johnson AP, Ross-Howe S, Peter E, Scott T, Ouellette C, et al. Postoperative remote automated monitoring: need for and state of the science. Can J Cardiol. 2018;34(7):850–62.

    Article  Google Scholar 

  2. de Vries EN, Ramrattan MA, Smorenburg SM, Gouma DJ, Boermeester MA. The incidence and nature of in-hospital adverse events: a systematic review. Qual Saf Health Care. 2008;17(3):216–23.

    Article  Google Scholar 

  3. Perman SM, Stanton E, Soar J, Berg RA, Donnino MW, Mikkelsen ME, Edelson DP, Churpek MM, Yang L, Merchant RM, et al. Location of in-hospital cardiac arrest in the United States-variability in event rate and outcomes. J Am Heart Assoc. 2016;5(10):e003638.

  4. Andersen LW, Berg KM, Chase M, Cocchi MN, Massaro J, Donnino MW. American Heart Association’s Get With The Guidelines-Resuscitation I: acute respiratory compromise on inpatient wards in the United States: incidence, outcomes, and factors associated with in-hospital mortality. Resuscitation. 2016;105:123–9.

    Article  Google Scholar 

  5. Pearse RM, Moreno RP, Bauer P, Pelosi P, Metnitz P, Spies C, Vallet B, Vincent JL, Hoeft A, Rhodes A, et al. Mortality after surgery in Europe: a 7 day cohort study. Lancet. 2012;380(9847):1059–65.

    Article  Google Scholar 

  6. Li G, Warner M, Lang BH, Huang L, Sun LS. Epidemiology of anesthesia-related mortality in the United States, 1999-2005. Anesthesiology. 2009;110(4):759–65.

    Article  Google Scholar 

  7. Morrison LJ, Neumar RW, Zimmerman JL, Link MS, Newby LK, McMullan PW Jr, Hoek TV, Halverson CC, Doering L, Peberdy MA, et al. Strategies for improving survival after in-hospital cardiac arrest in the United States: 2013 consensus recommendations: a consensus statement from the American Heart Association. Circulation. 2013;127(14):1538–63.

    Article  Google Scholar 

  8. Leuvan CH, Mitchell I. Missed opportunities? An observational study of vital sign measurements. Crit Care Resusc. 2008;10(2):111–5.

    PubMed  Google Scholar 

  9. Sun Z, Sessler DI, Dalton JE, Devereaux PJ, Shahinyan A, Naylor AJ, Hutcherson MT, Finnegan PS, Tandon V, Darvish-Kazem S, et al. Postoperative hypoxemia is common and persistent: a prospective blinded observational study. Anesth Analg. 2015;121(3):709–15.

    Article  CAS  Google Scholar 

  10. Duus CL, Aasvang EK, Olsen RM, Sorensen HBD, Jorgensen LN, Achiam MP, Meyhoff CS. Continuous vital sign monitoring after major abdominal surgery-quantification of micro events. Acta Anaesthesiol Scand. 2018;62(9):1200–8.

    Article  CAS  Google Scholar 

  11. Turan A, Chang C, Cohen B, Saasouh W, Essber H, Yang D, Ma C, Hovsepyan K, Khanna AK, Vitale J, et al. Incidence, severity, and detection of blood pressure perturbations after abdominal surgery: a prospective blinded observational study. Anesthesiology. 2019;130(4):550–9.

    Article  Google Scholar 

  12. Jones D, Mitchell I, Hillman K, Story D. Defining clinical deterioration. Resuscitation. 2013;84(8):1029–34.

    Article  Google Scholar 

  13. Lee LA, Caplan RA, Stephens LS, Posner KL, Terman GW, Voepel-Lewis T, Domino KB. Postoperative opioid-induced respiratory depression: a closed claims analysis. Anesthesiology. 2015;122(3):659–65.

    Article  CAS  Google Scholar 

  14. Schein RM, Hazday N, Pena M, Ruben BH, Sprung CL. Clinical antecedents to in-hospital cardiopulmonary arrest. Chest. 1990;98(6):1388–92.

    Article  CAS  Google Scholar 

  15. Franklin C, Mathew J. Developing strategies to prevent inhospital cardiac arrest: analyzing responses of physicians and nurses in the hours before the event. Crit Care Med. 1994;22(2):244–7.

    Article  CAS  Google Scholar 

  16. Chen L, Ogundele O, Clermont G, Hravnak M, Pinsky MR, Dubrawski AW. Dynamic and personalized risk forecast in step-down units. Implications for monitoring paradigms. Ann Am Thorac Soc. 2017;14(3):384–91.

    Article  Google Scholar 

  17. Saugel B, Cecconi M, Hajjar LA. Noninvasive cardiac output monitoring in cardiothoracic surgery patients: available methods and future directions. J Cardiothorac Vasc Anesth. 2019;33(6):1742–52.

    Article  Google Scholar 

  18. Saugel B, Cecconi M, Wagner JY, Reuter DA. Noninvasive continuous cardiac output monitoring in perioperative and intensive care medicine. Br J Anaesth. 2015;114(4):562–75.

    Article  CAS  Google Scholar 

  19. Saugel B, Khanna AK. Managing hemodynamic instability - if you want to know cardiac output, you need to measure it! J Crit Care. 2019;49:185–6.

    Article  Google Scholar 

  20. Bellomo R, Ackerman M, Bailey M, Beale R, Clancy G, Danesh V, Hvarfner A, Jimenez E, Konrad D, Lecardo M, et al. A controlled trial of electronic automated advisory vital signs monitoring in general hospital wards. Crit Care Med. 2012;40(8):2349–61.

    Article  Google Scholar 

  21. Subbe CP, Duller B, Bellomo R. Effect of an automated notification system for deteriorating ward patients on clinical outcomes. Crit Care. 2017;21(1):52.

    Article  Google Scholar 

  22. Taenzer AH, Pyke JB, McGrath SP, Blike GT. Impact of pulse oximetry surveillance on rescue events and intensive care unit transfers: a before-and-after concurrence study. Anesthesiology. 2010;112(2):282–7.

    Article  Google Scholar 

  23. Brown H, Terrence J, Vasquez P, Bates DW, Zimlichman E. Continuous monitoring in an inpatient medical-surgical unit: a controlled clinical trial. Am J Med. 2014;127(3):226–32.

    Article  Google Scholar 

  24. Michard F, Pinsky MR, Vincent JL. Intensive care medicine in 2050: NEWS for hemodynamic monitoring. Intensive Care Med. 2017;43(3):440–2.

    Article  Google Scholar 

  25. Michard F. A sneak peek into digital innovations and wearable sensors for cardiac monitoring. J Clin Monit Comput. 2017;31(2):253–9.

    Article  Google Scholar 

  26. Michard F. Hemodynamic monitoring in the era of digital health. Ann Intensive Care. 2016;6(1):15.

    Article  Google Scholar 

  27. Michard F, Sessler DI, Saugel B. Non-invasive arterial pressure monitoring revisited. Intensive Care Med. 2018;44(12):2213–5.

    Article  Google Scholar 

  28. Michard F, Sessler DI. Ward monitoring 3.0. Br J Anaesth. 2018;121(5):999–1001.

    Article  CAS  Google Scholar 

  29. Khanna AK, Overdyk FJ, Greening C, Di Stefano P, Buhre WF. Respiratory depression in low acuity hospital settings-seeking answers from the PRODIGY trial. J Crit Care. 2018;47:80–7.

    Article  Google Scholar 

  30. van Loon K, Peelen LM, van de Vlasakker EC, Kalkman CJ, van Wolfswinkel L, van Zaane B. Accuracy of remote continuous respiratory rate monitoring technologies intended for low care clinical settings: a prospective observational study. Can J Anaesth. 2018;65(12):1324–32.

    Article  Google Scholar 

  31. Weenk M, van Goor H, Frietman B, Engelen LJ, van Laarhoven CJ, Smit J, Bredie SJ, van de Belt TH. Continuous monitoring of vital signs using wearable devices on the general ward: pilot study. JMIR Mhealth Uhealth. 2017;5(7):e91.

    Article  Google Scholar 

  32. Michard F, Bellomo R, Taenzer A. The rise of ward monitoring: opportunities and challenges for critical care specialists. Intensive Care Med. 2018;45(5):671–73.

    Article  Google Scholar 

  33. Li X, Dunn J, Salins D, Zhou G, Zhou W, Schussler-Fiorenza Rose SM, Perelman D, Colbert E, Runge R, Rego S, et al. Digital health: tracking physiomes and activity using wearable biosensors reveals useful health-related information. PLoS Biol. 2017;15(1):e2001402.

    Article  Google Scholar 

  34. Chen L, Dubrawski A, Wang D, Fiterau M, Guillame-Bert M, Bose E, Kaynar AM, Wallace DJ, Guttendorf J, Clermont G, et al. Using supervised machine learning to classify real alerts and artifact in online multisignal vital sign monitoring data. Crit Care Med. 2016;44(7):e456–63.

    Article  Google Scholar 

  35. Kang MA, Churpek MM, Zadravecz FJ, Adhikari R, Twu NM, Edelson DP. Real-time risk prediction on the wards: a feasibility study. Crit Care Med. 2016;44(8):1468–73.

    Article  Google Scholar 

  36. Kovacs C, Jarvis SW, Prytherch DR, Meredith P, Schmidt PE, Briggs JS, Smith GB. Comparison of the National Early Warning Score in non-elective medical and surgical patients. Br J Surg. 2016;103(10):1385–93.

    Article  CAS  Google Scholar 

  37. Smith ME, Chiovaro JC, O'Neil M, Kansagara D, Quinones AR, Freeman M, Motu'apuaka ML, Slatore CG. Early warning system scores for clinical deterioration in hospitalized patients: a systematic review. Ann Am Thorac Soc. 2014;11(9):1454–65.

    Article  Google Scholar 

  38. Churpek MM, Yuen TC, Winslow C, Meltzer DO, Kattan MW, Edelson DP. Multicenter comparison of machine learning methods and conventional regression for predicting clinical deterioration on the wards. Crit Care Med. 2016;44(2):368–74.

    Article  Google Scholar 

  39. LeCun Y, Bengio Y, Hinton G. Deep learning. Nature. 2015;521(7553):436–44.

    Article  CAS  Google Scholar 

  40. McGillion M, Yost J, Turner A, Bender D, Scott T, Carroll S, Ritvo P, Peter E, Lamy A, Furze G, et al. Technology-enabled remote monitoring and self-management - vision for patient empowerment following cardiac and vascular surgery: user testing and randomized controlled trial protocol. JMIR Res Protoc. 2016;5(3):e149.

    Article  Google Scholar 

Download references

Acknowledgements

Not applicable

Funding

Not applicable

Author information

Authors and Affiliations

Authors

Contributions

AKK, PH, and BS performed the literature search, drafted the manuscript, created the figure, and approved the final version of the manuscript to be published.

Corresponding author

Correspondence to Bernd Saugel.

Ethics declarations

Ethics approval and consent to participate

Not applicable

Consent for publication

Not applicable

Competing interests

AKK collaborates with Medtronic (Boulder, CO, USA) as a member of the executive advisory board on respiratory monitoring and steering committee member of the PRODIGY trial and receives honoraria for these services including giving lectures and refunds of travels expenses. AKK serves on the clinical advisory board for Retia Medical (Valhalla, NY, USA), Linshom Medical (Ellicott City, MD, USA), and also serves as a consultant for La Jolla pharmaceuticals (San Diego, CA, USA) and as a subject matter expert for the development of the Anesthesia SimStat system for CAE healthcare (Sarasota, FL, USA). BS collaborates with Pulsion Medical Systems (Feldkirchen, Germany) as a member of the medical advisory board and received honoraria for giving lectures and refunds of travel expenses from Pulsion Medical Systems. BS received research support and honoraria for giving lectures from Edwards Lifesciences (Irvine, CA, USA). BS received institutional restricted research grants, honoraria for consulting, and refunds of travel expenses from Tensys Medical (San Diego, CA, USA). BS received honoraria for giving lectures and refunds of travel expenses from CNSystems Medizintechnik (Graz, Austria). BS received institutional restricted research grants from Retia Medical. BS received honoraria for giving lectures from Philips Medizin Systeme Böblingen (Böblingen, Germany). PH declared that he has no competing interests.

Additional information

Publisher’s Note

Springer Nature remains neutral with regard to jurisdictional claims in published maps and institutional affiliations.

Rights and permissions

Open Access This article is distributed under the terms of the Creative Commons Attribution 4.0 International License (http://creativecommons.org/licenses/by/4.0/), which permits unrestricted use, distribution, and reproduction in any medium, provided you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons license, and indicate if changes were made. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated.

Reprints and Permissions

About this article

Verify currency and authenticity via CrossMark

Cite this article

Khanna, A.K., Hoppe, P. & Saugel, B. Automated continuous noninvasive ward monitoring: future directions and challenges. Crit Care 23, 194 (2019). https://doi.org/10.1186/s13054-019-2485-7

Download citation

  • Received:

  • Accepted:

  • Published:

  • DOI: https://doi.org/10.1186/s13054-019-2485-7

Keywords

  • Hemodynamic monitoring
  • Postoperative complications
  • Blood pressure
  • Hypotension
  • Peripheral oxygen saturation
  • Hypoxemia
  • Remote monitoring
  • Normal ward
  • Artifacts
  • False alarms