Skip to main content

vvECMO can be avoided by a transpulmonary pressure guided open lung concept in patients with severe ARDS

Dear Editor,

The EOLIA trial concluded that vvECMO compared to conventional mechanical ventilation with low tidal volumes and airway pressures ≤30 cmH2O did not improve survival [1]. Although not statistically significant, the 11% absolute reduction in mortality rate and multiple crossovers to rescue vvECMO were considered to be clinically relevant [2]. However, a conventional mechanical ventilation strategy is likely to be insufficient for patients with severe ARDS, as higher airway pressures are required to maintain lung aeration [3]. Grasso et al. measured the transpulmonary pressure (PL) in patients with severe ARDS and increased PEEP until PL was 25 cmH2O. Fifty percent of patients responded to an increase in airway pressure and did not require vvECMO [4]. We hypothesized that a PL guided open lung concept (OLC) could improve oxygenation and prevent conversion to vvECMO in patients with severe ARDS.

We retrospectively reviewed the records of all patients referred to our ICU between January and May 2018. Eight patients had severe ARDS and had an indication for vvECMO according to the EOLIA trial (demographics are given in the Additional file 1) [1]. Before referral protective mechanical ventilation with low tidal volume and a plateau pressure of approximately 30 cmH2O was applied. PaO2/FiO2 ratio was 62 ± 7 mmHg despite the use of neuromuscular blocking agents and prone positioning. After referral, a recruitment maneuver was performed and PEEP was increased. PL was estimated with an esophageal balloon catheter and we aimed for a PL ≤ 25 cmH2O. In addition, respiratory rate and I:E ratio were increased, thereby generating intrinsic PEEP.

The PL guided OLC resulted in an increase in PaO2/FiO2 ratio to 201 ± 87 mmHg (Fig. 1) and none of the patients required vvECMO. During the first 6 h peak airway pressure was increased to 44.9 ± 10.2 cmH2O, but was reduced to 36.3 ± 5.6 cmH2O within 24 h, while PEEP was maintained at 20.6 ± 4.0 cmH2O. A maximum end-inspiratory PL of 18 ± 5 cmH2O was measured. At 72 h both peak airway pressures and PEEP were reduced to baseline values while oxygenation remained stable.

Fig. 1
figure 1

Airway pressures and PaO2 / FiO2 ratio after initiation of the OLC. Peak airway pressure, Pmean, PEEP and PaO2 / FiO2 ratio as a function of time. The OLC is initiated at T0, i.e. at referral. Mean values and standard deviations are shown. Note that PEEP values are set PEEP levels at the mechanical ventilator. The depicted driving pressure is overestimated as intrinsic PEEP is not shown. FiO2 fraction of inspired oxygen, PaO2 partial pressure of arterial oxygen, Ppeak peak airway pressure, Pmean mean airway pressure, PEEP positive end-expiratory pressure

These data suggest that the OLC improves oxygenation and avoids conversion to vvECMO in patients with severe ARDS. We acknowledge that a recruitment maneuver and higher PEEP in patients with moderate to severe ARDS increased mortality in the Alveolar Recruitment Trial [5]. However, the recruitment maneuver was standardized and ‘recruitability’ was not assessed. We hypothesize that a recruitment maneuver and higher PEEP is beneficial in patients with large regions of decreased lung aeration. Thus, future research should focus on individual ‘recruitability’ [6]. Baedorf Kassis et al. introduced a recruitment maneuver based on PL measurements [7]. Other potential predictors are a decrease in driving pressure, oxygenation response to PEEP-trials, or lung aeration estimated by electrical impedance tomography or ultrasound.

References

  1. Combes A, Hajage D, Capellier G, Demoule A, Lavoue S, Guervilly C, Da Silva D, Zafrani L, Tirot P, Veber B, Maury E, Levy B, Cohen Y, Richard C, Kalfon P, Bouadma L, Mehdaoui H, Beduneau G, Lebreton G, Brochard L, Ferguson ND, Fan E, Slutsky AS, Brodie D, Mercat A, Eolia Trial Group R, Ecmonet. Extracorporeal membrane oxygenation for severe acute respiratory distress syndrome. N Engl J Med. 2018;378:1965–75.

    Article  Google Scholar 

  2. Gattinoni L, Vasques F, Quintel M. Use of ECMO in ARDS: does the EOLIA trial really help? Crit Care. 2018;22:171.

    Article  Google Scholar 

  3. van der Zee P, Gommers D, (2019) Recruitment maneuvers and higher PEEP, the so-called open lung concept, in patients with ARDS. Crit Care 23: 73.

  4. Grasso S, Terragni P, Birocco A, Urbino R, Del Sorbo L, Filippini C, Mascia L, Pesenti A, Zangrillo A, Gattinoni L, Ranieri VM. ECMO criteria for influenza a (H1N1)-associated ARDS: role of transpulmonary pressure. Intensive Care Med. 2012;38:395–403.

    Article  CAS  Google Scholar 

  5. Writing Group for the Alveolar Recruitment for Acute Respiratory Distress Syndrome Trial I, Cavalcanti AB, Suzumura EA, Laranjeira LN, Paisani DM, Damiani LP, Guimaraes HP, Romano ER, Regenga MM, Taniguchi LNT, Teixeira C, Pinheiro de Oliveira R, Machado FR, Diaz-Quijano FA, Filho MSA, Maia IS, Caser EB, Filho WO, Borges MC, Martins PA, Matsui M, Ospina-Tascon GA, Giancursi TS, Giraldo-Ramirez ND, Vieira SRR, Assef M, Hasan MS, Szczeklik W, Rios F, Amato MBP, Berwanger O, Ribeiro de Carvalho CR. Effect of lung recruitment and titrated positive end-expiratory pressure (PEEP) vs low PEEP on mortality in patients with acute respiratory distress syndrome: a randomized clinical trial. Jama. 2017;318:1335–45.

    Article  Google Scholar 

  6. Chiumello D, Brochard L, Marini JJ, Slutsky AS, Mancebo J, Ranieri VM, Thompson BT, Papazian L, Schultz MJ, Amato M, Gattinoni L, Mercat A, Pesenti A, Talmor D, Vincent JL. Respiratory support in patients with acute respiratory distress syndrome: an expert opinion. Crit Care. 2017;21:240.

    Article  Google Scholar 

  7. Baedorf Kassis E, Loring SH, Talmor D. Recruitment maneuvers: using transpulmonary pressure to help goldilocks. Intensive Care Med. 2017;43:1162–3.

    Article  CAS  Google Scholar 

Download references

Funding

None.

Availability of data and materials

The dataset used and/or analyzed during the current study are available from the corresponding author on reasonable request.

Author information

Authors and Affiliations

Authors

Contributions

PZ drafted the manuscript, DM, JM, HE, and DG substantially revised the manuscript. All authors approved the submitted version.

Corresponding author

Correspondence to Philip van der Zee.

Ethics declarations

Ethics approval and consent to participate

This retrospective study was approved by the medical ethics committee of the Erasmus MC (MEC-2018-1300). According to Dutch law no informed consent was required with anonymous retrospective data.

Consent for publication

Not applicable.

Competing interests

Dinis Dos Reis Miranda received speakers fee and travel expenses from Xenios and Hill-Rom.

Diederik Gommers received speakers fee and travel expenses from Dräger, GE Healthcare (medical advisory board 2009–2012), Maquet, and Novalung (medical advisory board).

Philip van der Zee, Han Meeder, and Henrik Endeman report no competing interests.

Publisher’s Note

Springer Nature remains neutral with regard to jurisdictional claims in published maps and institutional affiliations.

Additional file

Additional file 1:

Figure S1. Flowchart of patient inclusion. Table S1. Patient demographics. Table S2. Patient parameters. Appendix Mechanical ventilation strategy. (DOCX 38 kb)

Rights and permissions

Open Access This article is distributed under the terms of the Creative Commons Attribution 4.0 International License (http://creativecommons.org/licenses/by/4.0/), which permits unrestricted use, distribution, and reproduction in any medium, provided you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons license, and indicate if changes were made. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated.

Reprints and Permissions

About this article

Verify currency and authenticity via CrossMark

Cite this article

van der Zee, P., Dos Reis Miranda, D., Meeder, H. et al. vvECMO can be avoided by a transpulmonary pressure guided open lung concept in patients with severe ARDS. Crit Care 23, 133 (2019). https://doi.org/10.1186/s13054-019-2421-x

Download citation

  • Received:

  • Accepted:

  • Published:

  • DOI: https://doi.org/10.1186/s13054-019-2421-x