Early detection of acute kidney injury (AKI) is challenging due to the risk of morbidity and mortality and a direct impact on patients’ management [1]. The diagnosis relies on the changes of serum creatinine and urine output [2], which are the main markers of kidney function. Recently, Astute Medical introduced the NephroCheck, a test that allows a bedside analysis of two biomarkers of renal damage implicated in G1 cell-cycle arrest: tissue inhibitor metalloproteinase-2 (TIMP-2) and insulin-like growth factor binding protein-7 (IGFBP-7) [3]. The combination of these two biomarkers led to a new score (AKIRisk™). An AKIRisk™ score > 0.3 identifies patients at risk of developing AKI with sensitivity and specificity of 92% and 46%, respectively; increasing the cutoff to 2.0, the sensitivity is 46% and the specificity is 95% [4]. The AKIRisk™ reference interval in healthy humans ranges from 0.04 to 2.22. A possible reason for this wide range could be that the score is not taking into account urine concentration.
We aimed to check the correlation between AKIRisk™ and urine osmolality, using a dehydration test. We collected urine samples from healthy volunteers after 8 h of operating room shift without drinking water (T0) and after drinking 0.5 l of water (T1). Urine samples were analyzed, and osmolality as well as biomarker concentration were measured. Complete measurements are reported in Additional file 1: Table S1. A significant difference was found between the mean AKIRisk™ at T0 (0.82, 95% CI 0.15 to 1.48) vs. T1 (0.24, 95% CI 0.02 to 0.50), p = 0.01 (Wilcoxon test—Fig. 1a). The Pearson correlation between osmolality and AKIRisk™ at T0 and T1 was r = 0.93, p = 0.02, and r = 0.80, p = 0.03 (Fig. 1b, c).
Our results suggest that fluid intake in the normal population is able to modify the urinary concentration of TIMP-2 and IGFB-7. It is to note that every AKIRisk™ > 0.3 occurs in people with urine osmolality > 600 mOsm/kg and that there is a good correlation between urine osmolality and AKIRisk™. Some participants still maintained AKIRisk™ > 0.3 even after fluid reintegration, maintaining a good correlation with urinary osmolality also at T1, indicating a suboptimal dehydration correction.
Our data suggest that the values of AKIRisk™ score could be related to the urine concentration; thus, urine osmolality should be considered in the interpretation of the results of the NephroCheck® test. This correlation should be checked in critically ill patients at risk of AKI.
References
Martin C, Cortegiani A, Gregoretti C, Martin-Loeches I, Ichai C, Leone M et al. Choice of fluids in critically ill patients. BMC Anesthesiology. 2018;18:200.
Kidney Disease: Improving Global Outcomes (KDIGO) Acute Kidney Injury Work Group. KDIGO clinical practice guideline for acute kidney injury. Kidney Inter. 2012;2:1–138.
Kashani K, Al-Khafaji A, Ardiles T, et al. Discovery and validation of cell cycle arrest biomarkers in human acute kidney injury. Crit Care. 2013;17:R25.
Bihorac A, Chawla LS, Shaw AD, Al-Khafaji A, Davison DL, Demuth GE, et al. Validation of cell-cycle arrest biomarkers for acute kidney injury using clinical adjudication. Am J Respir Crit Care Med. 2014;189:932–9.
Acknowledgements
None.
Funding
None.
Availability of data and materials
Not applicable.
Author information
Authors and Affiliations
Contributions
AN, AC, and AD conceived the content and wrote and approved the final version of the manuscript.
Corresponding author
Ethics declarations
Ethics approval and consent to participate
Not applicable.
Consent for publication
Not applicable.
Competing interests
The authors declare that they have no competing interests.
Publisher’s Note
Springer Nature remains neutral with regard to jurisdictional claims in published maps and institutional affiliations.
Additional file
Additional file 1:
Raw data. Urine osmolality and AKIRisk score of each patient. (DOCX 43 kb)
Rights and permissions
Open Access This article is distributed under the terms of the Creative Commons Attribution 4.0 International License (http://creativecommons.org/licenses/by/4.0/), which permits unrestricted use, distribution, and reproduction in any medium, provided you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons license, and indicate if changes were made. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated.
About this article
Cite this article
Noto, A., Cortegiani, A. & David, A. NephroCheck: should we consider urine osmolality?. Crit Care 23, 48 (2019). https://doi.org/10.1186/s13054-019-2341-9
Received:
Accepted:
Published:
DOI: https://doi.org/10.1186/s13054-019-2341-9