- Research
- Open access
- Published:
Detection of inspiratory recruitment of atelectasis by automated lung sound analysis as compared to four-dimensional computed tomography in a porcine lung injury model
Critical Care volume 22, Article number: 50 (2018)
Abstract
Background
Cyclic recruitment and de-recruitment of atelectasis (c-R/D) is a contributor to ventilator-induced lung injury (VILI). Bedside detection of this dynamic process could improve ventilator management. This study investigated the potential of automated lung sound analysis to detect c-R/D as compared to four-dimensional computed tomography (4DCT).
Methods
In ten piglets (25 ± 2 kg), acoustic measurements from 34 thoracic piezoelectric sensors (Meditron ASA, Norway) were performed, time synchronized to 4DCT scans, at positive end-expiratory pressures of 0, 5, 10, and 15 cmH2O during mechanical ventilation, before and after induction of c-R/D by surfactant washout. 4DCT was post-processed for within-breath variation in atelectatic volume (Δ atelectasis) as a measure of c-R/D. Sound waveforms were evaluated for: 1) dynamic crackle energy (dCE): filtered crackle sounds (600–700 Hz); 2) fast Fourier transform area (FFT area): spectral content above 500 Hz in frequency and above −70 dB in amplitude in proportion to the total amount of sound above −70 dB amplitude; and 3) dynamic spectral coherence (dSC): variation in acoustical homogeneity over time. Parameters were analyzed for global, nondependent, central, and dependent lung areas.
Results
In healthy lungs, negligible values of Δ atelectasis, dCE, and FFT area occurred. In lavage lung injury, the novel dCE parameter showed the best correlation to Δ atelectasis in dependent lung areas (R2 = 0.88) where c-R/D took place. dCE was superior to FFT area analysis for each lung region examined. The analysis of dSC could predict the lung regions where c-R/D originated.
Conclusions
c-R/D is associated with the occurrence of fine crackle sounds as demonstrated by dCE analysis. Standardized computer-assisted analysis of dCE and dSC seems to be a promising method for depicting c-R/D.
Background
Although positive pressure ventilation can be life-saving by restoring adequate oxygenation, mechanical ventilation itself can lead to secondary lung damage [1, 2]. In addition to volutrauma and barotrauma, atelectrauma (cyclic recruitment and de-recruitment of atelectasis, or c-R/D) also contributes to ventilator-induced lung injury (VILI) [2, 3].
Numerous studies have investigated c-R/D and addressed the specific role of atelectrauma in experimental settings. Using dynamic computed tomography (dCT), within-breath recruitment and de-recruitment were visualized by variations in atelectatic lung fractions [4, 5]. Further experimental studies demonstrated that c-R/D leads to respiration-dependent oscillations in blood oxygenation that originate in the lungs [6] and are forwarded downstream via the circulation to the end-organ level [7, 8]. In this context, more severe lung tissue damage and an increased inflammatory response have been shown in lung areas where c-R/D occurs [9, 10], highlighting the relevance of c-R/D to the onset of VILI.
Recently, several novel ventilatory strategies have been proposed for the purpose of avoiding c-R/D during mechanical ventilation [11,12,13], In clinical practice, however, bedside detection of the dynamic process of c-R/D is not possible with currently available tools.
A noninvasive, bedside method that might be adapted for the detection of c-R/D is automated lung sound auscultation [14, 15]. The first attempt to assess tidal recruitment by automated lung sound analysis was presented by Vena and colleagues. They post-processed an acoustic parameter that reflects the changes in spectral characteristics of lung sounds during inspiration [16], termed “fast Fourier transform area” (FFT area).
Our study focused on the technical development of a novel sound-based parameter for the detection of within-breath recruitment, in a model where the within-breath changes in atelectasis (Δ atelectasis) could be verified by the reference method of four-dimensional computed tomography (4DCT). For our investigations, we proposed to induce a broad range of c-R/D conditions by setting different positive end-expiratory pressure (PEEP) levels at a fixed end-inspiratory pressure level of 30 cmH2O, resulting in different tidal volumes. In this setup, we aimed to capture and quantify the distinct sound signature (i.e., adventitious sounds) associated with the sudden opening of atelectatic lung units during inspiration by post-processing the “dynamic crackle energy” (dCE) in the frequency range of 600 to 700 Hz. Moreover, we aimed to localize the origin of c-R/D acoustically by assessing the “dynamic changes in spectral coherence” (dSC) throughout inspiration.
As such, we hypothesized that there is a linear correlation between Δ atelectasis and dCE, and Δ atelectasis and the reproduced FFT area parameter, respectively. Additionally, we hypothesized that dSC is different in regard to different lung regions and PEEP levels.
Methods
Animal experiments
Following Animal Care Committee approval (Landesuntersuchungsamt Koblenz) of the Rhineland Palatinate, Germany (23,177-07/G09-1-029), 10 piglets were studied. One animal was needed to set up the protocol. Two animals expired during c-R/D induction and one did not provide a complete dataset due to technical failures. Thus, six animals were included in the final analysis. All procedures were performed under deep anesthesia, and careful efforts were made to minimize suffering.
After induction of general anesthesia, catheters (for the purposes of invasive monitoring) were surgically placed. Details concerning the anesthetic procedures and routine monitoring regimen can be found in Additional file 1.
Preliminary experimental tests
Before carrying out the study, we assessed the influence that surrounding noise might exert upon the attached piezoelectric contact sensors, and the influence the sensors themselves might exert upon the radiologic imaging quality. Using a noise-absorbing mat that we wrapped around the subjects, pretesting showed that the recorded raw data sound waveforms were not noticeably affected by external noise. Furthermore, no specific artifacts could be attributed to the sensor positioning. Concerning the metallic acoustic sensors themselves, we found that they produced a bias of up to 40 Hounsfield Units (HU) on mean lung densities (MLDs) in computed tomography (CT) imaging (see Additional file 1: Figure S1).
Automated lung sound recordings and four-dimensional computed tomography
In our experimental setup, 36 piezoelectric contact sensors (Meditron ASA, Oslo, Norway) were arranged into two arrays (one on the left side and one on the right), with each array containing 18 sensors organized into three columns (M1–M3) and six rows (L1–L6). Two of the 36 sensors were inactive and served as a reference for ambient noise. Figure 1 demonstrates the anatomical sensor positions. The sensor arrays were placed in a circular fashion around the pig’s thorax using special gel pads, thus forming apical (columns M3 and M5), middle (columns M2 and M6), and basal (columns M1 and M7) transversal sensor planes. Correspondingly, the first and second rows of sensors (L1 and L2) covered the nondependent lung areas, the third and fourth rows (L3 and L4) covered the central lung areas, and the fifth and six rows (L5 and L6) covered the dependent lung areas. Placement of the sensor arrays was CT-guided so as to position the basal transversal sensor plane 2.5 cm above the dome of the diaphragm. Based on the predefined sensor matrix, the sensors of the middle transversal sensor plane were located between the orifice of the upper right and middle right lung lobes, while the apical transversal sensor plane was located roughly at the bifurcation of the trachea. The sensor arrays were connected to the vibration response imaging (VRI) device (VRIxv, GE Healthcare, Little Chalfont, UK) and raw data acoustic waveforms were collected at a sampling frequency of 19,200 Hz.
4DCT measurements (Brilliance iCT 256-slice scanner, Philips, Amsterdam, the Netherlands) were performed on identical lung regions with a cranio-caudal span of 8 cm, correlating directly with the placement of the sensor array matrix. In accordance with the anatomical positions of the acoustic sensors, nondependent, central, and dependent lung regions were analyzed.
Study protocol
Measurements of lung sound acoustics were performed which were time-synchronized to 4DCT at randomly set PEEP levels of 0, 5, 10, and 15 cmH2O during both healthy baseline (BLH) and after induction of the model lung injury (surfactant depletion injury (LAV)).
To study c-R/D, surfactant depletion was induced via repetitive lung lavages using isotonic solution (30 ml/kg) until reaching a lung state characterized by substantial lung collapse (defined as a Horowitz-index < 300 at zero end-expiratory pressure (ZEEP)), but still retaining the capability of partial within-breath recruitment (defined as Horowitz-index < 450 at a PEEP of 15 cmH2O). This model was similar to that used in one of our previously published studies [13]. A pressure controlled ventilation (PCV) regimen was chosen. To produce a broad range of c-R/D conditions, we used different PEEP levels (so as to vary static recruitment) at a fixed end-inspiratory pressure of 30 cmH2O, and an inspiration-to-expiration ratio of 1:1. This resulted in different driving pressures and different tidal volumes for the investigation of within-breath recruitment, while keeping the mechanism of recruitment unchanged.
Each PEEP level was maintained for at least 10 min; then, data were recorded for a period of 20 s at a respiratory rate of 6 breaths/min due to the limited temporal resolution of the CT scanner.
Offline data handling of four-dimensional computed tomography scans
Similar to the methodology utilized in a previous study [17], quantitative analysis of CT attenuation of lung tissue was carried out semi-automatically using an in-house-developed software (YACTA version 1.09.40, University of Mainz, Germany), which was written by one of the authors (OW). Details about 4DCT post-processing are provided in the supplemental section (Additional file 1: Figure S2). The behavior of atelectatic (−300 to 0 HU), poorly aerated (−600 to −301 HU), normally aerated (−900 to −601 HU), and hyperinflated (−1024 to −901 HU) lung volumes were computed over the time course of the breathing cycle in steps of 0.58 s. The amount of c-R/D was evaluated by assessing the differences between end-expiratory and end-inspiratory values in the atelectatic lung volume (Δ atelectasis). An example of 4DCT post-processing appears in the supplemental section (Additional file 1: Figure S3).
Offline data handling of automated lung sound recordings: overview
Raw data sound waveforms were post-processed to evaluate three different parameters.
The first parameter was dynamic crackle energy (dCE). This parameter reflects the amount of sound energy in the frequency spectrum of 600 to 700 Hz over the inspiratory time course of the breathing cycle [14, 18, 19]. According to the literature, this is the defined frequency band of fine crackle sounds [20,21,22].
For the second parameter, we reproduced the fast Fourier transform (FFT) analysis introduced by Vena et al. [16] (termed FFT area) by calculating the spectral content above 500 Hz in frequency and above −70 dB in amplitude in proportion to the total amount of sound above −70 dB amplitude.
The third parameter was spectral coherence (SC)/dynamic spectral coherence (dSC). These parameters reflect regional acoustical homogeneity in the subjacent lung regions of neighboring acoustic sensors (SC) and their variation over the inspiratory time course (dSC) [23].
Essentially, the first two parameters were used to search for a linear correlation between Δ atelectasis and dCE, and Δ atelectasis and FFT area, while the purpose of the dSC parameter was to localize the origin of within-breath recruitment acoustically.
An example of the data handling of acoustic lung sounds is given in Fig. 2.
Details of the automated lung sound analysis
All analyses were primarily carried out for each acoustic sensor and in time clips of 0.58 s to match the demand acquisition time of 4DCT scanning, resulting in paired measurements over the entire breathing cycle. For final statistics, results were subsumed for the inspiratory phase of the entire (global) lung, and regionally for dependent, central, and nondependent lung areas, by summarizing the parameter values of the respective acoustic sensors overlaying the defined lung regions (see Fig. 1). All computing was performed using the programming environment MATLAB and Simulink Toolbox Release 2014b (The MathWorks, Inc., Natick, MA, USA).
Assessment of the dynamic crackle energy (dCE)
The dCE parameter was computed as follows. Raw data sound waveforms were downsampled by a factor of 4 from 19,200 Hz to 4800 Hz. Then, fine crackle sounds were isolated using a band-pass filter with finite impulse response (FIR; two-hundredth order with the cut-off frequencies of 600 and 700 Hz). In a subsequent step, for each audio clip of 0.58 s, the root mean squares (RMS) of downsampled sound amplitudes were calculated.
Assessment of the spectral characteristics of lung sounds (FFT area)
For the assessment of the FFT area, the following operations were performed. After downsampling to 4800 Hz, the signals were bandpass filtered (FIR-filter of two hundredth order in the frequency range of 75–2000 Hz). For each audio clip of 0.58 s, the power spectral densities of the signals were computed (using Welch’s method to convert the signals from the time to the frequency domain by Fourier transform: 200 samples window length, 50% overlap, Hamming-windowing) and transformed to decibels (dB). Using the resulting graph (Additional file 1: Figure S4), the area under the curve (AUC) above −70 dB in amplitude and above 500 Hz in frequency was related back to the AUC above −70 dB and expressed as a percentage.
Assessment of regional spectral coherence (SC) and its variation over time (dSC)
To localize the origin of c-R/D acoustically, the spectral coherence method was used (representing a function of frequency that indicates how well two sounds match at each frequency, i.e., the better the match, the better the homogeneity of sounds from adjacent lung regions). Raw data waveforms were downsampled (4800 Hz) and bandpass filtered (75–2000 Hz). Then, the signals of neighboring acoustic sensors were analyzed in regards to their mutual spectral coherence. For each lung region of interest (ROI), the arithmetic mean of all pairs of acoustic sensors overlaying the predefined lung regions was assessed in time clips of 0.58 s. From the resulting spectral coherence time plot, two parameters were computed: the spectral coherence (SC) by averaging all values over the inspiratory phase, and the dynamic spectral coherence (dSC) by assessing the time-dependent variation over the inspiratory phase of the breathing cycle (Additional file 1: Figure S5). A more detailed description of how SC and dSC were computed is available in Additional file 1.
Statistics
The relationships between Δ atelectasis by 4DCT and dCE and FFT area, respectively, were analyzed using linear mixed models (LMMs). These were fitted for the entire region of interest (global), and regionally for nondependent, central, and dependent lung areas (dCT as nondependent variable; dCE and FFT area as dependent variables; piglet ID as random intercept to account for the structure of dependency due to repeated measures; Bonferroni-Holm method for multiple testing). Based on the model intercepts and slopes, the corresponding regression lines and the marginal R2 were computed [24]. Differences in dSC in regard to lung region (nondependent, central, dependent) and in regard to PEEP (0 and 15 cmH2O), were addressed by another LMM, which also tested the interaction between lung region and PEEP. We accounted for the structure of dependency due to repeated measures (Piglet ID as random intercept) and adjusted for multiple testing by the Bonferroni-Holm method.
For descriptive statistics, mean and standard deviation values are reported. Statistics were performed using the statistical software R (R: A Language and Environment for Statistical Computing, R Core Team, R Foundation for Statistical Computing, Vienna, Austria), GraphPad Prism v6 (GraphPad Software Inc., San Diego, CA, USA), and Origin (OriginLab, Northampton, MA, USA).
Results
Results of healthy baseline measurements
Under healthy conditions, c-R/D was not evident on 4DCT in any of the animals. BLH measurements found a negligible amount of atelectasis (ranging from 1.6 to 18.2 cm3) with Δ atelectasis ranging from 0 to 13.3 cm3 (ZEEP to PEEP, 15 cmH2O). The total volume of the lung stack examined by 4DCT was 295 ± 24 cm3, yielding a volume percentage change in atelectasis ranging from 0 to 4.5% of the lung volume imaged. In the synchronous recorded sound waveforms, post-processed values for dCE were minimal, ranging from 0.019 to 0.041, and for FFT area from 3.6 to 13.8% for all PEEP steps and lung regions. Overall, under baseline conditions, the analysis of SC showed high values (56 ± 4.7), with minimal changes over the time-course of inspiration, resulting in dSC values of 2.2 ± 0.5.
Results of model lung injury measurements
Induction of lung injury by 3 ± 1 lavages induced c-R/D in all subjects, as confirmed by 4DCT. Ventilatory, gas exchange, and hemodynamic parameters are presented in Table 1. Routinely, the highest Δ atelectasis occurred at ZEEP. Within-breath changes in atelectasis decreased with increasing PEEP at the predefined ventilator settings. The variation of PEEP levels (0, 5, 10, and 15 cmH2O) at a fixed end-inspiratory pressure of 30 cmH2O induced a large range of Δ atelectasis values (from 5.8 to 60 cm3). This represented a lung volume change in atelectasis ranging from 1.8 to 20.3% of the imaged lung volume. Table 2 summarizes the atelectatic lung volumes and their changes over the respiratory cycle for the defined lung regions. The full set of post-processed lung volumes are presented in Additional file 1: Figure S6, Tables S1 and S2.
Automated lung sound analysis found that adventitious sounds in the frequency range between 600 and 700 Hz (dCE) occurred synchronous to a shift in sound spectral characteristics above −70 dB in amplitude and above 500 Hz in frequency (FFT area) in the presence of c-R/D. Both dCE and FFT area exhibited the highest values at ZEEP, which were reduced when Δ atelectasis decreased as PEEP was increased. For the entire lung region, and for each subregion examined, the dCE and the FFT area varied with Δ atelectasis, with higher correlations for each lung region for the dCE analysis. The detailed results of the LMM analyses with regard to the level of PEEP, as well as the analyzed region, are shown in Fig. 3.
Interestingly, dCE and FFT area signals predominately arose in the first 1 to 2 s after the initiation of inspiration, a time period when 4DCT indicated the greatest changes in atelectatic lung volume (Fig. 4). The full dataset is presented in Additional file 1: Figure S7.
The LMM analysis of spectral coherence found that PEEP (P = 0.0031) and lung ROI (P = 0.0002) had a significant influence on dSC, whereas their interaction (PEEP × ROI) was not significant (P = 0.2633). Estimates with standard error of the pair-wise comparisons of the fitted analysis of variance model are presented in Additional file 1: Table S3. We found that dSC significantly differed between dependent and nondependent lung regions (P < 0.0001), as well as between dependent and central lung regions (P = 0.0067). No significant effect was found between central and nondependent lung regions (P = 0.0696). As investigated by variation of PEEP, dSC values decreased from 3.8 ± 0.6/5.7 ± 2.4/8.1 ± 2.4 (nondependent/central/dependent, respectively) at ZEEP to 3.2 ± 0.4/3.9 ± 1.4/5.3 ± 2.4 at a PEEP of 15 cmH2O, while SC values increased from 33.3 ± 1.4/36.5 ± 4.2/37.6 ± 4.2 at ZEEP to 42.9 ± 4.1/49.3 ± 4.4/54.5 ± 9.6 at a PEEP of 15 cmH2O (Additional file 1: Figure S8).
Discussion
The present study assessed the potential of automated lung sound analysis for quantifying inspiratory recruitment of atelectasis during mechanical ventilation. Computer-assisted analysis of lung sound recordings could potentially provide continuous, noninvasive, bedside detection of c-R/D with no known hazards (e.g., exposure to ionizing radiation). We used an experimental model of lung lavage to produce surfactant depletion and facilitate c-R/D [13], and we used differing levels of PEEP to vary the amount of c-R/D over a broad range. 4DCT, covering a thick axial lung segment, was used as a standard to define the amount of tidal recruitment of atelectasis. We used one previously reported method [16], as well as two new methods, for analyzing lung sounds to identify and localize within-breath recruitment of atelectasis. Our study showed promising correlations between these sound analysis methods and the amount of tidal recruitment, as assessed by within-breath variation in atelectatic volume by the reference method of 4DCT.
Since the introduction of the stethoscope by Laennec [25], clinicians have used qualitative sound analysis as a diagnostic tool to identify lung pathologies. More recently, automated recording systems have become available that can capture, store, and analyze lung sounds quantitatively to provide further information [26, 27]. Automated lung sound analysis is noninvasive, observer-independent, and allows for an objective measurement and classification of acoustic pattern at standardized conditions. Despite this, its application in a clinical setting is limited due to multiple sources of ambient noise that might bias the results, even though electronic auscultation has the advantage of signal amplification and ambient noise reduction. Although we used a commercially available multisensing technology to record acoustic waveforms, our evaluations were performed on a raw data level, focusing on the detection of adventitious sounds associated with c-R/D. We used the VRI system simply as a methodological tool to standardize lung recordings via a proven, state-of-the-art piezoelectric multisensor recording technology. The idea that c-R/D may generate a distinct sound signature was based on prior studies that attributed fine crackles—defined as short, nonmusical, explosive sounds, typically hovering around a frequency of 650 Hz [18]—to airway opening [19]. Thus, we aimed to capture and analyze sound waves in an experimental model where cyclical recruitment of atelectasis was documented by 4DCT as the reference method.
Our data showed that c-R/D was associated with the occurrence of adventitious crackle sounds during mechanical ventilation. The novel dCE parameter presented here could quantify these crackles and correlated well to within-breath variation in atelectatic volume as assessed by 4DCT. The best dCE results (R2 = 0.88) were found by the acoustic sensors overlaying the dependent lung regions where c-R/D originated. Additionally, the FFT area analysis (reproduced from Vena et al. [16]) was also well suited to detecting c-R/D when analyzing all acoustic sensors (R2 = 0.72), but was less useful for regional discrimination (R2 = 0.31–0.56). Overall, less tight correlations by FFT area (R2 = 0.31–0.72) were found when compared to dCE (R2 = 0.56–0.88).
To identify the regions of the lung where c-R/D was taking place, we computed the time-dependent variation of spectral coherence of neighboring acoustic sensors during inspiration. Our data showed that highest dSC values occurred in the dependent lung regions and at ZEEP, which was in agreement with 4DCT results, showing that the phenomenon of c-R/D took place at the functional border of atelectatic and poorly ventilated lung compartments. Additionally, increasing the PEEP led to a significant reduction in dSC, while absolute SC values increased, which might be best explained by the restoration of lung homogeneity.
Our study also suggests that acoustic methods can provide some insights into the kinetics of recruitment. The variation in time of dCE and FFT area analysis suggests that the majority of intratidal recruitment took place in the first 1 to 2 s, in agreement with the time course of changes in atelectasis as assessed by 4DCT. Similar time constants during inflation have been reported previously [5, 13, 28].
One of our study’s strengths was the use of 4DCT with a longitudinal coverage of 8 cm as the reference standard. Additionally, the multisensor sound recording system allowed for regional subanalyses. The 4DCT and acoustic methods were precisely time-aligned for all measurements. The study design included large tidal volumes and randomly varying PEEP levels that allowed for observations over a wide range of c-R/D. Our ventilation regimen, however, was not intended to mimic a clinical scenario, but to experimentally induce a large range of c-R/D. Additionally, the acquisition time for 4DCT restricted our study to very slow respiratory rates, although from a technical point of view the dCE parameter could be processed at any physiological respiratory rate. Considering this, we used sharp borders to quantify the amount of cyclic atelectasis, and since the metallic sensors per se had an influence on CT attenuation, the chosen HU range might in part include poorly aerated lung tissue. Moreover, we cannot exclude the possibility that the noise present in the CT signal (which may have been due to these sensors) and the interference due to movement during mechanical ventilation might have biased the weights of the results and statistics. Our study is an evaluation of the potential that sound analysis has to monitor cyclical recruitment of atelectasis, and as such it was carried out in a carefully defined laboratory setting with minimal surrounding noise. Many issues would need to be addressed in order to translate this potential into bedside clinical practice. Although we do not as of yet claim a direct clinical application for our results, this does not preclude consideration of the basic mechanistic concepts for eventual clinical use, possibly even a solution which combines multiparametric lung sound analysis with other noninvasive bedside technologies for the assessment of lung function (e.g., electrical impedance tomography). Concurrent pathophysiologic processes could generate competing lung sounds or alter sound transmission in ways that obscure the distinct sound signatures of inspiratory recruitment, e.g., bronchospasm, fibrotic lung changes, and pneumonia [29,30,31,32,33]. Moreover, the present work cannot define to what extent changes in tidal volume bias the detection of the presented dCE parameter. Thus, one could claim that the reported decrease in Δ atelectasis could be the effect of a decrease in tidal volume rather than a decrease in cyclic lung units opening and closing. We cannot completely exclude this possibility due to our study design (which used varying tidal volumes instead of constant tidal volumes). We believe that this is unlikely, as the FFT area analysis is not dependent upon tidal volume, and both FFT area and dCE yielded high linear correlations for the entire lung stack throughout the various PEEP levels and tidal volumes. Finally, as can be appreciated from Fig. 1, current technology for multisensory sound detection is a bit cumbersome for use in the clinical setting in mechanically ventilated patients, although the technology has been successfully applied in other clinical settings. Although the initial parameter calculations were performed offline—which admittedly was quite time-consuming—the necessary post-processing steps were then converted into a fully automated MATLAB routine that is suitable for integration into any automated lung sound device. Using this script running on a computer with MATLAB, the lung sound parameter calculations take mere seconds, do not require any operator intervention, and provide for the possibility of real-time analysis. The next step would certainly be the implementation of those algorithms via software updates into current automated lung sound devices. Concerning the practicality of lung sound analysis in routine clinical use, we acknowledge that the relevant technology is in need of further development in its current form. Current systems, some of them having been in development for decades now, are not yet streamlined or “simplified” enough for the clinical bedside setting. Nonetheless, these technologies must be continually revisited and enhanced with the latest technological developments (i.e., noise canceling, sensor miniaturization, systems integration, etc.) so as to eventually produce devices and procedures that may be of everyday clinical use, beyond what is currently considered to be possible.
Conclusions
c-R/D is associated with the occurrence of fine crackle sounds as demonstrated by dCE analysis. Standardized computer-assisted analysis of dCE, in combination with dSC analysis, seems to be a promising method for depicting c-R/D, as shown in an experimental model of surfactant-depleted pigs. Overall, this method was found to be superior to FFT area analysis. The novel parameters presented here for the purpose of acoustical quantification of c-R/D, however, warrant and require further clinical study under realistic conditions before these experimental findings might be translated into clinical practice.
Abbreviations
- 4DCT:
-
Four-dimensional computed tomography
- AUC:
-
Area under the curve
- BLH:
-
Healthy baseline conditions
- c-R/D:
-
Cyclic recruitment and de-recruitment of atelectasis
- CT:
-
Computed tomography
- dCE:
-
Dynamic crackle energy
- dCT:
-
Dynamic computed tomography
- dSC:
-
Dynamic spectral coherence
- FFT:
-
Fast Fourier transform
- FIR:
-
Finite impulse response
- HU:
-
Hounsfield Units
- LAV:
-
Surfactant depletion injury
- LMM:
-
Linear mixed model
- PEEP:
-
Positive end-expiratory pressure
- ROI:
-
Region of interest
- SC:
-
Spectral coherence
- VILI:
-
Ventilator-induced lung injury
- VRI:
-
Vibration response imaging
- ZEEP:
-
Zero end-expiratory pressure
References
Dreyfuss D, Saumon G. Ventilator-induced lung injury. Am J Respir Crit Care Med. 1998;157:294–323.
Slutsky AS, Ranieri VM. Ventilator-induced lung injury. N Engl J Med. 2013;369:2126–36.
Caironi P, Cressoni M, Chiumello D, Ranieri M, Quintel M, Russo SG, et al. Lung opening and closing during ventilation of acute respiratory distress syndrome. Am J Respir Crit Care Med. 2010;181:578–86.
David M, Karmrodt J, Bletz C, David S, Herweling A, Kauczor H-U, et al. Analysis of atelectasis, ventilated, and hyperinflated lung during mechanical ventilation by dynamic CT. Chest. 2005;128:3757–70.
Markstaller K, Eberle B, Kauczor H-U, Scholz A, Bink A, Thelen M, et al. Temporal dynamics of lung aeration determined by dynamic CT in a porcine model of ARDS. Br J Anaesth. 2001;87:459–68.
Baumgardner JE, Markstaller K, Pfeiffer B, Doebrich M, Otto CM. Effects of respiratory rate, plateau pressure, and positive end-expiratory pressure on PaO2 oscillations after saline Lavage. Am J Respir Crit Care Med. 2002;166:1556–62.
Klein KU, Hartmann EK, Boehme S, Szczyrba M, Heylen L, Liu T, et al. PaO2 oscillations caused by cyclic alveolar recruitment can be monitored in pig buccal mucosa microcirculation. Acta Anaesthesiol Scand. 2013;57:320–5.
Klein KU, Boehme S, Hartmann EK, Szczyrba M, Heylen L, Liu T, et al. Transmission of arterial oxygen partial pressure oscillations to the cerebral microcirculation in a porcine model of acute lung injury caused by cyclic recruitment and derecruitment. Br J Anaesth. 2013;110:266–73.
Otto CM, Markstaller K, Kajikawa O, Karmrodt J, Syring RS, Pfeiffer B, et al. Spatial and temporal heterogeneity of ventilator-associated lung injury after surfactant depletion. J Appl Physiol (1985). 2008;104:1485–94.
Sinclair SE, Chi E, Lin H-I, Altemeier WA. Positive end-expiratory pressure alters the severity and spatial heterogeneity of ventilator-induced lung injury: an argument for cyclical airway collapse. J Crit Care. 2009;24:206–11.
Hartmann EK, Boehme S, Bentley A, Duenges B, Klein KU, Elsaesser A, et al. Influence of respiratory rate and end-expiratory pressure variation on cyclic alveolar recruitment in an experimental lung injury model. Crit Care. 2012;16:R8.
Syring RS, Otto CM, Spivack RE, Markstaller K, Baumgardner JE. Maintenance of end-expiratory recruitment with increased respiratory rate after saline-lavage lung injury. J Appl Physiol (1985). 2007;102:331–9.
Boehme S, Bentley AH, Hartmann EK, Chang S, Erdoes G, Prinzing A, et al. Influence of inspiration to expiration ratio on cyclic recruitment and derecruitment of atelectasis in a saline lavage model of acute respiratory distress syndrome. Crit Care Med. 2015;43:e65–74.
Bohadana A, Izbicki G, Kraman SS. Fundamentals of lung auscultation. N Engl J Med. 2014;370:744–51.
Dellinger RP, Jean S, Cinel I, Tay C, Rajanala S, Glickman YA, et al. Regional distribution of acoustic-based lung vibration as a function of mechanical ventilation mode. Crit Care. 2007;11:R26.
Vena A, Rylander C, Perchiazzi G, Giuliani R, Hedenstierna G. Lung sound analysis correlates to injury and recruitment as identified by computed tomography: an experimental study. Intensive Care Med. 2011;37:1378–83.
Ley-Zaporozhan J, Ley S, Unterhinninghofen R, Weinheimer O, Saito Y, Kauczor H-U, et al. Quantification of lung volume at different tidal volumes and positive end-expiratory pressures in a porcine model by using retrospective respiratory gated 4D-computed tomography. Investig Radiol. 2008;43:461–9.
Munakata M, Ukita H, Doi I, Ohtsuka Y, Masaki Y, Homma Y, et al. Spectral and waveform characteristics of fine and coarse crackles. Thorax. 1991;46:651–7.
Vyshedskiy A, Alhashem RM, Paciej R, Ebril M, Rudman I, Fredberg JJ, et al. Mechanism of inspiratory and expiratory crackles. Chest. 2009;135:156–64.
Robertson A, Coope R. Râles, Rhonchi, and Laennec. Lancet. 1957;273:417–23.
Forgacs P. The functional basis of pulmonary sounds. Chest. 1978;73:399–405.
Mikami R, Murao M, Cugell DW, Chrétien J, Cole P, Meier-Sydow J, et al. International symposium on lung sounds. Chest. 1987;92:342–5.
Kay SM. Modern spectral estimation: theory and application. Englewood Cliffs: Prentice Hall; 1988.
Nakagawa S, Schielzeth H. A general and simple method for obtaining R2 from generalized linear mixed-effects models. Methods Ecol Evol. 2012;4:133–42.
Laennec RTH. De lauscultation mediate ou traité du diagnostic des maladies des Poumons et du Cæur, fondé principalement sur ce nouveau moyen dexploration. N Engl J Med. 1821;10:132–56.
Dellinger RP, Parrillo JE, Kushnir A, Rossi M, Kushnir I. Dynamic visualization of lung sounds with a vibration response device: a case series. Respiration. 2008;75:60–72.
Shi C, Boehme S, Bentley AH, Hartmann EK, Klein KU, Bodenstein M, et al. Assessment of regional ventilation distribution: comparison of vibration response imaging (VRI) with electrical impedance tomography (EIT). PLoS One. 2014;9:e86638.
Albert SP, Dirocco J, Allen GB, Bates JHT, Lafollette R, Kubiak BD, et al. The role of time and pressure on alveolar recruitment. J Appl Physiol (1985). 2009;106:757–65.
Cottin V, Richeldi L. Neglected evidence in idiopathic pulmonary fibrosis and the importance of early diagnosis and treatment. Eur Respir Rev. 2014;23:106–10.
Murphy RL, Vyshedskiy A, Power V-A, Bana D, Marinelli P, Wong-Tse A, et al. Automated lung sound analysis in patients with pneumonia. Respir Care. 2004;49:1490–7.
al Jarad N, Strickland B, Bothamley G, Lock S, Logan-Sinclair R, Rudd RM. Diagnosis of asbestosis by a time expanded wave form analysis, auscultation and high resolution computed tomography: a comparative study. Thorax. 1993;48:347–53.
Nath AR, Capel LH. Lung crackles in bronchiectasis. Thorax. 1980;35:694–9.
Piirilä P, Sovijärvi AR, Kaisla T, Rajala HM, Katila T. Crackles in patients with fibrosing alveolitis, bronchiectasis, COPD, and heart failure. Chest. 1991;99:1076–83.
Acknowledgments
We would like to thank the Department of Anesthesiology, Medical Center of the Johannes-Gutenberg University Mainz, Mainz, Germany, for the provision of facilities and equipment.
Funding
The project was funded by the German Research Council (Deutsche Forschungsgemeinschaft) grant number DFG Pak 415: Ma 2398/6. The VRI device was provided by G.E. Healthcare Inc. for research purposes.
Availability of data and materials
All data analyzed during this study are included in this published article and its supplementary information files. The raw datasets used for the analysis are available from the corresponding author on reasonable request.
Author information
Authors and Affiliations
Contributions
SB takes responsibility for the content of the manuscript, was involved in the conception, hypotheses delineation, and design of the study, acquisition and analysis of the data, and in writing the article. EKH, AHB, and YY were involved in the design of the study, the conduct of the experiments, and the revision of the manuscript prior to submission. FPRT, EK, OW, and TA were involved in the analysis of the data and in its revision prior to submission. MH reviewed the raw data, was responsible for statistical analysis and figure preparation, and was involved in writing the article and in the revision of this article prior to submission. JEB and KM were involved in the conception, hypotheses delineation, and design of the study, and revised this article prior to submission. All authors approved the final version of the manuscript.
Corresponding author
Ethics declarations
Ethics approval
Animal Care Committee approval, Landesuntersuchungsamt Koblenz, Germany: 23,177-07/G09-1-029.
Consent for publication
Not applicable.
Competing interests
The authors declare that they have no competing interests.
Publisher’s Note
Springer Nature remains neutral with regard to jurisdictional claims in published maps and institutional affiliations.
Additional file
Additional file 1:
Supplemental material and supporting information (containing the supplemental Figure S1–S9 and the supplemental Tables S1–S3). (PDF 8554 kb)
Rights and permissions
Open Access This article is distributed under the terms of the Creative Commons Attribution 4.0 International License (http://creativecommons.org/licenses/by/4.0/), which permits unrestricted use, distribution, and reproduction in any medium, provided you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons license, and indicate if changes were made. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated.
About this article
Cite this article
Boehme, S., Toemboel, F.P.R., Hartmann, E.K. et al. Detection of inspiratory recruitment of atelectasis by automated lung sound analysis as compared to four-dimensional computed tomography in a porcine lung injury model. Crit Care 22, 50 (2018). https://doi.org/10.1186/s13054-018-1964-6
Received:
Accepted:
Published:
DOI: https://doi.org/10.1186/s13054-018-1964-6