With much interest we read the paper written by Saugel et al. [1] on a systemic approach to ultrasound-guided central vein catheterization. The article is comprehensive but some issues need discussing.

The authors claim that the tip of the needle can be constantly identified while the needle is approaching the vein both in short-axis and long-axis views. This is inaccurate. The only approach where it is possible to constantly visualize the tip of the needle is the in-plane technique. In the short-axis view, the needle is visible as a white dot, which also applies to the tip of the needle as with any part of the needle shaft [2, 3]. It means that the tip can be within the lumen of the vein or below the vessel when the second wall of the vein is punctured.

The scheme created by the authors is similar to the one we recently published [4]. Unfortunately, their proposed systematic approach ignores several crucial steps in the procedure. Firstly, the needle is not always visualized perfectly. Therefore, before puncturing the vein, you need to check if the needle has not been positioned over the adjacent artery. How this is done depends on the technique used. It is simple in the short-axis view, when both vessels are constantly presented on the screen, but in the long-axis view, only the vein is visualized; thus, the angle needs to be changed between the ultrasound probe and the skin in order to visualize the adjacent artery. When the tip of the needle is not centrally positioned over the vein, the needle can miss the vein, also damaging the adjacent artery.

The other important step is to test the introduction of the guidewire, i.e., to check whether the tip of the guidewire is not being introduced into the other central vein on the same side of the patient. This problem mostly concerns the jugular vein during axillary/subclavian vein catheterization and the axillary/subclavian vein during jugular vein catheterization. This step acts as a protection (but not completely) against introducing the catheter into an undesirable location.

In our opinion ultrasound-guided central venous catheter placement is a slightly more complex procedure than the one proposed in the paper. It not only involves introducing the catheter into the vein but also controls all the steps in order to guarantee the safety of the patient.

References

  1. Saugel B, Scheeren TWL, Teboul JL. Ultrasound-guided central venous catheter placement: a structured review and recommendations for clinical practice. Crit Care. 2017;21(1):225.

    Article  PubMed  PubMed Central  Google Scholar 

  2. Reusz G, Csomos A. The role of ultrasound guidance for vascular access. Curr Opin Anaesthesiol. 2015;28(6):710–6.

    PubMed  Google Scholar 

  3. Labib A, Bodenham A. Ultrasound-guided vascular access: The basics. In: Lumb P, Karakitsos D, editors. Critical care ultrasound. Philadelphia: Elsevier; 2015. p. 66–74.

    Google Scholar 

  4. Gawda R, Czarnik T, Łysenko L. Infraclavicular access to the axillary vein—the new possibilities for the catheterization of the central veins in the intensive care unit. Anaesthesiol Intensive Ther. 2016;48(5):360–6.

    Article  PubMed  Google Scholar 

Download references

Acknowledgments

None.

Funding

None.

Availability of data and materials

Not applicable.

Author information

Authors and Affiliations

Authors

Contributions

RG wrote the manuscript. TC revised the manuscript. Both authors read and approved the final manuscript.

Corresponding author

Correspondence to Ryszard Gawda.

Ethics declarations

Ethics approval and consent to participate

Not applicable.

Consent for publication

RG and TC consent for publication.

Competing interests

The authors declare that they have no competing interests.

Publisher’s Note

Springer Nature remains neutral with regard to jurisdictional claims in published maps and institutional affiliations.

Rights and permissions

Open Access This article is distributed under the terms of the Creative Commons Attribution 4.0 International License (http://creativecommons.org/licenses/by/4.0/), which permits unrestricted use, distribution, and reproduction in any medium, provided you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons license, and indicate if changes were made. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated.

Reprints and permissions

About this article

Check for updates. Verify currency and authenticity via CrossMark

Cite this article

Gawda, R., Czarnik, T. A systematic approach to ultrasound-guided central venous catheter placement—desirable modifications. Crit Care 21, 299 (2017). https://doi.org/10.1186/s13054-017-1877-9

Download citation

  • Received:

  • Accepted:

  • Published:

  • DOI: https://doi.org/10.1186/s13054-017-1877-9