VCO2 calorimetry is a convenient method for improved assessment of energy expenditure in the intensive care unit

The Original Article was published on 21 March 2016

In their interesting article, Stapel et al. [1] suggested the use of carbon dioxide production (VCO2) calorimetry with energy expenditure (EE; kcal/day), calculated as 8.19 × VCO2 (ml/min), where VCO2 is provided by the built-in capnometer of the mechanical ventilator. This calculation on average overestimated EE by 7.7 % compared to indirect calorimetry (IC) with a standard deviation (SD) of ±8.4 %. This is within the ±10 % limits of acceptance used in many studies [2] and, more importantly, is an improvement relative to calculation of EE by predictive equations from the patient’s anthropometric data. The equation used by Stapel et al. incorporated a cohort respiratory quotient (RQ) of 0.86.

In his commentary, Pierre Singer [3] questions the usability of VCO2 for assessing EE in critically ill patients for three reasons: 1) the concept involves complicated mathematics; 2) calculation of RQ from the patient’s nutrition is complicated and uncertain; and 3) this invalidates the use of VCO2 calorimetry in the critically ill patient. We disagree with the first reason. Multiplying VCO2 by 8.19 is not complicated. We agree with the second reason. We unexpectedly saw significantly lower RQs in patients on a glucose-only diet compared with patients on enteral nutrition, such that individual RQ estimates calculated from the nutrition would have been inaccurate [4]. This does not imply that we agree with the third reason. In our sensitivity analysis we showed that changing our mean cohort RQ of 0.81 to 0.76, which is the lower end of the published range, only increased the VCO2 calorimetry estimates of EE by 6 %, while increasing RQ to the upper end of the published range, RQ = 0.89, reduced estimated EE by 8 %.

We recommended choosing a value of RQ = 0.85. With that choice, VCO2 calorimetry on average underestimated EE by 4 %, with an SD of 3 %, relative to EE estimated by IC [4], well within the ±10 % limits of acceptance. Our findings agree well with those by Stapel et al. and our conclusion is that VCO2 calorimetry is both easy and usable as a method for assessing EE for any cohort RQ within the published range (0.76–0.89).

The question remains whether VCO2 measured by built-in capnographs in various ventilators is sufficiently accurate. Stapel et al. found a 6.6 % systematic overestimation of VCO2 with their ventilator (SERVO-i; Maquet), compared to the gold standard (Deltatrac II; Datex). This is promising, but data are needed for other built-in capnographs.

Author’s response

We would like to thank Pielmeier and Andreassen for their valuable arguments in response to Pierre Singer’s commentary on our study about VCO2-derived measurement of EE.

In this study we concluded that EE can be accurately assessed at the bedside by multiplying ventilator-derived VCO2 by 8.19, especially when taking the mean 24-h value [1]. To calculate EE from ventilator-derived VCO2, we had to estimate the RQ in order to transform VCO2 into VO2. For study purpose, we used the RQ of nutritional intake, knowing there would be inaccuracy [5]. We actually found that measured RQ (0.859 ± 0.047) was quite similar to nutritional RQ (0.864 ± 0.015), but the two were not related at all (p = 0.485). In addition, the SD of the measured RQ was much larger. Thus, we share the opinion of both Pielmeier and Singer that calculating RQ from the administered nutrition is inaccurate. Nutritional RQ does not account for endogenous metabolism, which is unpredictable during critical illness. This was reported previously by McClave et al. [5] and is confirmed by the study of Rousing et al. [4], to which Pielmeier and Andreassen contributed.

Nevertheless, as we outlined, RQ only accounts for 15 % of the total bias of VCO2-derived EE. Since the mean RQ of our cohort and the RQ of most nutritional formulae are both 0.86, we proposed the simplified equation based on a fixed RQ of 0.86: EE = 8.19 × VCO2 (ml/min). In their interesting study, Rousing et al. [4] showed that, for any chosen RQ within the range of cohort values of 0.76 to 0.89, VCO2-based calorimetry performed significantly better than equations, thereby confirming our findings. They recommend using an RQ of 0.85, surprisingly similar to our conclusion.

Of note, IC remains the gold standard for assessment of EE in ventilated critically ill patients. However, the best-validated system, the Deltatrac, is no longer on the market and new indirect calorimeters have not yet proven to be accurate [6]. More importantly, predictive equations are inaccurate and their use should be avoided. VCO2-based EE provides the best alternative.

We agree with Pielmeier and Andreassen that we cannot extrapolate the results of our capnograph to other built-in capnographs. Their accuracy should first be validated, especially during irregular breathing. Furthermore, since we found a systematic error of 6.6 % for the VCO2 measurement, accuracy of the Maquet measurement should also be improved. The use of VCO2 measurements per second (instead of per minute) is currently under investigation.

Abbreviations

EE, energy expenditure; IC, indirect calorimetry; RQ, respiratory quotient; SD, standard deviation; VCO2, carbon dioxide production

References

  1. Stapel SN, de Grooth HJS, Alimohamad H, Elbers PW, Girbes AR, Weijs PJ, et al. Ventilator-derived carbon dioxide production to assess energy expenditure in critically ill patients: proof of concept. Crit Care. 2015;19:370.

    Article  PubMed  PubMed Central  Google Scholar 

  2. Tatucu-Babet OA, Ridley EJ, Tierney AC. Prevalence of underprescription or overprescription of energy needs in critically ill mechanically ventilated adults as determined by indirect calorimetry: a systematic literature review. J Parenter Enteral Nutr. 2016;40(2):212–25.

    Article  CAS  Google Scholar 

  3. Singer P. Simple equations for complex physiology: can we use VCO2 for calculating energy expenditure? Crit Care. 2016;20:72.

    Article  PubMed  PubMed Central  Google Scholar 

  4. Rousing ML, Hahn Pedersen MH, Andreassen S, Pielmeier U, Preiser JC. Energy expenditure in critically ill patients estimated by population based equations, indirect calorimetry and CO2-based indirect calorimetry. Ann Intensive Care. 2016;6(1):16.

    Article  PubMed  PubMed Central  Google Scholar 

  5. McClave SA, Lowen CC, Kleber MJ, McConnell JW, et al. Clinical use of the respiratory quotient obtained from indirect calorimetry. J Parenter Enteral Nutr. 2003;27(1):21–6.

    Article  Google Scholar 

  6. Sundstrom Rehal M, Fiskaare E, Tjader I, et al. Measuring energy expenditure in the intensive care unit: a comparison of indirect calorimetry by E-sCOVX and Quark RMR with Deltatrac II in mechanically ventilated critically ill patients. Crit Care. 2016;20:54.

    Article  Google Scholar 

Download references

Authors’ contributions

UP and SA drafted the manuscript. Both authors approved the final manuscript.

Competing interests

The authors declare that they have no competing interests.

Author information

Authors and Affiliations

Authors

Corresponding author

Correspondence to Ulrike Pielmeier.

Additional information

See related article by Stapel et al., http://ccforum.biomedcentral.com/articles/10.1186/s13054-015-1087-2, and related article by Singer, http://ccforum.biomedcentral.com/articles/10.1186/s13054-016-1251-3

Rights and permissions

Open Access This article is distributed under the terms of the Creative Commons Attribution 4.0 International License (http://creativecommons.org/licenses/by/4.0/), which permits unrestricted use, distribution, and reproduction in any medium, provided you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons license, and indicate if changes were made. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated.

Reprints and permissions

About this article

Check for updates. Verify currency and authenticity via CrossMark

Cite this article

Pielmeier, U., Andreassen, S. VCO2 calorimetry is a convenient method for improved assessment of energy expenditure in the intensive care unit. Crit Care 20, 224 (2016). https://doi.org/10.1186/s13054-016-1397-z

Download citation

  • Received:

  • Accepted:

  • Published:

  • DOI: https://doi.org/10.1186/s13054-016-1397-z

Keywords