The relationship between arterial transducer level and pulse contour waveform-derived measurements

For hemodynamic monitoring, the pressure transducer is suggested to be fixed at the level of the phlebostatic axis in critically ill patients [1,2]. The correction and adjustment of pressure transducer are emphasized in central venous pressure monitoring in clinical practice. The exact position of the transducer is relatively easy to be ignored for invasive arterial blood pressure monitoring [3,4]. Improper position of the transducer may cause inaccurate value and shape of the arterial blood pressure wave, which would result in an invalid PiCCO (Pulsion Medical Systems AG, Munich, Germany) algorithm for pulse contour waveform-derived measurements. This study was conducted as a prospective quantitative evaluation of the relationship between arterial transducer level and pulse contour waveform-derived measurements.

In total, 22 patients were enrolled in the 28-bed department of critical care medicine of a university hospital. All of the patients had a femoral artery catheter for PiCCO hemodynamic monitoring. The site of the phlebostatic axis was defined as the zero level (reference level). We moved the arterial pressure transducer up and down at eight different levels (−5 cm, −10 cm, −15 cm, −20 cm, 5 cm, 10 cm, 15 cm, 20 cm). At each level, continuous cardiac index (CCI), rate of left ventricular pressure rise during systole (dP/dtmax), and systemic vascular resistance index (SVRI) were simultaneously recorded.

The elevation of pressure transducer caused significantly positive changes in CCI and negative changes in dP/dtmax and SVRI, which resulted in a change in the opposite direction for these parameters (−CCI, +dP/dtmax, and + SVRI). When the change of the transducer’s position was 5 cm, the changes of SVRI with error reached statistical significance (P < 0.0001), but there were no differences in the CCI between the 5 cm and 0 cm reference level (CCI: +5 cm versus 0 cm 3.2 ± 0.6 versus 3.2 ± 0.7, P = 0.715; −5 cm versus 0 cm 3.1 ± 0.7 versus 3.2 ± 0.7, P = 0.075). When the variation of transducer level was 10 cm, the change of CCI with error reached statistical significance (P < 0.0001) (Figure 1). Different positions of the transducer and the corresponding changes in CCI and dP/dtmax are shown in Table 1. Figure 2 shows the trends of estimated marginal means of CCI, CCI change rate, dP/dtmax, and dP/dtmax change rate at different positions of the transducer.

Figure 1
figure 1

The variation of continuous cardiac index (CCI) and systemic vascular resistance index (SVRI) at different levels of the pressure transducer. *P < 0.0001 versus 0 cm transducer position. CCI is presented as liters per minute per square meter, and SVRI is presented as dyne per second per centimeter raised to the fifth power.

Table 1 Corresponding vertical distances and the changes and variation rates of continuous cardiac index changes and dP/dtmax
Figure 2
figure 2

The variation of estimated marginal means of continuous cardiac index (CCI), CCI change rate, rate of left ventricular pressure rise during systole (dP/dtmax), and dP/dtmax change rate.

The contour waveform-derived parameter response to the change transducer is still undetermined. To the best of our knowledge, this is the first study to explore the impact of arterial transducer level on pulse contour waveform-derived parameters in clinical practice. The clinical implications of such errors are important to recognize. Our study provided evidence that the arterial transducer should be considered in order to obtain precise pulse contour waveform-derived parameters, especially when the transducer’s vertical distance was more than 10 cm from the phlebostatic axis. We believe that these findings deserve emphasis and should be applied in continuous hemodynamic monitoring.

Abbreviations

CCI:

Continuous cardiac index

dP/dtmax:

rate of left ventricular pressure rise during systole

SVRI:

Systemic vascular resistance index

References

  1. Rauen CA, Makic MB, Bridges E. Practice evidence-based practice habits: transforming research into bedside. Crit Care Nurse. 2009;29:46–59.

    Article  Google Scholar 

  2. Winsor T, Burch G. Phlebostatic level: reference level for venous pressure measurement in man. Proc Soc Exp Biol Med. 1945;58:165–9.

    Article  Google Scholar 

  3. McCann 2nd UG, Schiller HJ, Carney DE, Kilpatrick J, Gatto LA, Paskanik AM. Invasive arterial BP monitoring in trauma and critical care. Chest. 2001;120:1322–6.

    Article  Google Scholar 

  4. He HW, Liu DW. Passive leg raising: influence of blood pressure transducer site. Intensive Care Med. 2013;39:1668.

    Article  Google Scholar 

Download references

Acknowledgments

The Institutional Research and Ethics Committee of the Peking Union Medical College Hospital approved this study for human subjects. Written informed consent was obtained from all patients or next of kin before data were included in the study. This work was supported by the special fund for health-scientific research in the public interest program (grant number 201202011, the Ministry of Health of P.R. China). The authors thank for Jing Zhang for reviewing the manuscript for spelling and grammar.

Author information

Authors and Affiliations

Authors

Corresponding author

Correspondence to Dawei Liu.

Additional information

Competing interests

The authors declare that they have no competing interests.

Rights and permissions

This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.

Reprints and permissions

About this article

Check for updates. Verify currency and authenticity via CrossMark

Cite this article

He, H., Liu, D., Long, Y. et al. The relationship between arterial transducer level and pulse contour waveform-derived measurements. Crit Care 19, 31 (2015). https://doi.org/10.1186/s13054-015-0745-8