Comparison of five sedation scoring systems by means of auditory evoked potentials

Comments

The drawbacks of AERs are that they require training and experience, and the effects of repeated auditory stimuli on a critically ill patient are unknown. The newer bispectral index monitor may be more useful in the ICU setting since it is simple to use by all staff and requires no auditory input. Alternatively in the absence of these expensive monitoring systems, oversedation may be avoided by switching off sedation at regular periods and not restarting infusions until the patient reaches a lesser sedation score.

Introduction

Auditory evoked responses (AEP) have been used as a means of monitoring awareness during anaesthesia. Attention has focused on the latency of the early cortical response Nb which indicates light anaesthesia and potential awareness when decreased to less than 44.5 ms. On the intensive care unit (ICU) sedation and analgesia is necessary to prevent ventilator dysynchrony and allow unpleasant procedures to take place. Sedation scoring systems are usually used to assess the level of sedation but unfortunately they are subjective and poor indicators of oversedation.

Aims

The authors have compared an objective measure (AER) with various subjective scoring systems.

Methods

The study investigated 95 consecutive patients requiring sedation for ventilation on the ICU. Midazolam and fentanyl were given for sedation and analgesia respectively, but additional agents could be used if sedation proved inadequate (propofol, methohexitone, gamma-hydroxybutyrate, ketamine or clonidine). The AER (midlatency Nbpeak) and corresponding level of sedation was recorded using five scoring systems [Ramsay, Cohen, Cambridge, Bloomsbury and Newcastle (Cook) Sedation Scores]. Kendall's correlation coefficients were calculated and regression analysis performed.

Results

The authors carried out 190 measurements in the 95 patients. Patients with higher APACHE II scores were more likely to be deeply sedated since, for example, severe acute respiratory distress syndrome (ARDS) patients required greater sedation to prevent ventilator dysynchrony and impaired gas exchange.The Ramsay score revealed the best correlation with changes in Nb latency (see table)

Table

However scatterplots showed a large variation of Nb latencies with deeper levels of sedation; ie these scoring systems are poor discriminators at deeper sedation levels.

Discussion

Derived electrophysiological measurements and evoked potentials have all been used to measure depth of sedation/anaesthesia but require training and experience to develop competency. However, since clinical scoring systems are poor discriminators at deeper sedation levels, the authors suggest that AERs may have a role where deeper levels of sedation are required for prolonged periods. It is postulated that the Ramsay score showed the best correlation since it includes a primitive reflex (glabellar tap) to discriminate asleep states.

References

  1. Schulte-Tamburen AM, Scheier J, Briegel J, Schwender D, Peter K: Comparison of five sedation scoring systems by means of auditory evoked potentials. Intensive Care Med. 1999, 25: 377-382.

    Article  PubMed  CAS  Google Scholar 

Download references

Authors

Rights and permissions

Reprints and permissions

About this article

Cite this article

Venn, R. Comparison of five sedation scoring systems by means of auditory evoked potentials . Crit Care 1, 121 (1999). https://doi.org/10.1186/ccf-1999-121

Download citation

  • Published:

  • DOI: https://doi.org/10.1186/ccf-1999-121

Keywords