Accuracy of a continuous non-invasive hemoglobin monitor in the ICU

Introduction

The non-invasive pulse CO-oximeter provides an immediate and continuous estimation of hemoglobin concentration non-invasively, and so has the potential to improve ICU patient care. We determined whether non-invasive hemoglobin measurement by pulse CO-oximetry could provide clinically acceptable absolute and trend accuracy in critically ill patients, compared with other invasive methods of hemoglobin assessment available at bedside and the gold standard, the laboratory analyzer.

Methods

A prospective study was conducted in the surgical ICU of a university teaching hospital. Blood samples from subjects continuously monitored with pulse CO-oximetry (SpHb) were analyzed for hemoglobin concentration determination by a point-of-care device (HemoCue301, HbHC), satellite laboratory CO-oximetry (Siemens RapidPoint 450, HbABG) and a laboratory hematology analyzer (Sysmex XT-2000i, tHB), which was considered the reference device. Hemoglobin values reported from the invasive methods were compared with the values reported by the Masimo Radical-7 Pulse CO-Oximeter at the time of the blood draw.

Results

Sixty-two patients requiring 471 blood samples were included. Compared with the reference method, the bias and limits of agreement were 0.0 ± 1.0 g/dl for SpHb, 0.3 ± 1.3 g/dl for HbHC and 0.9 ± 0.6 g/dl for HbABG compared with the reference device. Accuracy assessed with ARMS was 0.8 g/dl for SpHb and 1.1 g/dl for HbABG and HbHC. Pulse CO-oximetry showed similar trend accuracy as CO-oximetry, whereas the point-of-care device did not follow the trend of the laboratory device as well as the other analyzers. See Table 1.

Table 1 Accuracy summary of test devices compared with laboratory hematology analyzer

Conclusions

When compared with laboratory reference values, hemoglobin measurement with pulse CO-oximetry has absolute and trending accuracy similar to widely used, invasive methods such as CO-oximetry and a point-of-care device. Hemoglobin measurement with pulse CO-oximetry has the additional advantages of providing continuous measurements, non-invasively, which may facilitate hemoglobin monitoring in the ICU.

Author information

Authors and Affiliations

Authors

Rights and permissions

This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.

Reprints and permissions

About this article

Cite this article

Frasca, D., Dahyot-Fizelier, C., Karen, C. et al. Accuracy of a continuous non-invasive hemoglobin monitor in the ICU. Crit Care 15 (Suppl 1), P429 (2011). https://doi.org/10.1186/cc9849

Download citation

  • Published:

  • DOI: https://doi.org/10.1186/cc9849

Keywords