Accuracy of glucose measurements in critically ill patients

Introduction

The aim as to compare the accuracy of glucose measurements using a glucose meter analyser infingerstick, arterial blood and laboratory tests in critically ill patients and to determine factors influencing the bias between these methods.

Methods

This prospective observational study included 75 consecutive ICU patients, corresponding to 302 pairs of measures (aged 56 ± 16 years, SOFA 8 ± 4, IGS2 56 ± 20). Findings from two different methods of glucose measurement were compared with laboratory blood glucose measurements: glucose meter analysis of capillary blood (fingerstick), and glucose meter analysis of arterial blood during the first 3 days in the ICU. Agreement between measurements was assessed using the Bland-Altman method.

Results

The correlation coefficient between fingerstick and laboratory results was 0.91 (95% CI = 0.89 to 0.99) and 0.92 (95% CI = 0.92 to 0.94) between arterial glucose meter analysis and laboratory testing. The mean bias between fingerstick and laboratory testing was 0.16 g/l, and between arterial glucose meter and laboratory testing was 0.10 g/l. Norepinephrine used did not affect the bias between methods (P = NS for all comparisons). See Figure 1.

Figure 1
figure 1

(abstract P406).

Conclusions

The agreement between methods for glucose level measurements appears not to be clinically acceptable. Either fingerstick or arterial glucose meter analysis have large limits of agreement with the gold standard laboratory testing for blood glucose measurements.

Author information

Authors and Affiliations

Authors

Rights and permissions

This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.

Reprints and permissions

About this article

Cite this article

Claverie, V., Lonjaret, L., Riu-Poulenc, B. et al. Accuracy of glucose measurements in critically ill patients. Crit Care 15 (Suppl 1), P406 (2011). https://doi.org/10.1186/cc9826

Download citation

  • Published:

  • DOI: https://doi.org/10.1186/cc9826

Keywords