Clinical experiences with a new endobronchial blocking device: the EZ-Blocker

Introduction

Both elective and emergency thoracic surgical procedures may require one-lung ventilation (OLV) for lung isolation [1]. Although in the majority of the cases a double lumen endotracheal tube (DLT) is the first choice, there are situations when insertion of DLT is not feasible [2]. We therefore intended to test the applicability of a recently developed endobronchial blocker (BB), the EZ-Blocker, in clinical practice.

Methods

Data were obtained from 10 patients undergoing thoracic surgery necessitating OLV. For lung isolation, a single lumen tube (SLT) and EZ-Blocker as BB were used. The time of insertion and positioning of BB, the lung deflation time with the BB cuff inflated and deflated, the minimal occlusion volume (MOV) of the BB cuff with 25 cmH2O positive airway pressure (PAP) and intracuff pressure (ICP) at MOV were registered. Based on the CT scan the diameter of the right (RMB) and left main bronchus (LMB) at 1 cm distal apart from the bifurcation was measured offline. Lung deflation was defined as 5.5 cm distance of the upper lobe from the rib cage at open chest.

Results

The insertion time was 76 ± 15 seconds. The lung deflation time through the lumen with the BB cuff inflated was 700 ± 83 seconds, and with a deflated cuff through the lumen of SLT was 9.4 ± 0.7 seconds. The MOV was 6.7 ± 1 ml in LMB versus 8 ± 1 ml in RMB (P = 0.03). The ICP was 40 ± 4 mmHg in LMB versus 85 ± 5 mmHg in RMB (P < 0.001). With linear regression there were strong positive relationships between the diameter of MB and MOV/ICP.

Conclusions

The use of EZ-Blocker is easy and safe for infrequent users, too. The short insertion time and short lung deflation time allows use in an emergency situation or in case of a difficult airway. Only a small fraction of ICP (10 to 20%) is transmitted to the bronchial wall and it does not cause mucosal ischemia. The diameter of the MB has great impact on the MOV and ICP. The MOV is similar but ICP is smaller than published in previous reports with other BBs [3].

References

  1. Mungroop HE, et al.: Br J Anaesth. 2010, 104: 119-120. 10.1093/bja/aep353

    Article  CAS  PubMed  Google Scholar 

  2. Benumof JL: J Cardiothor Vasc Anesth. 1998, 12: 131-132. 10.1016/S1053-0770(98)90317-2

    Article  CAS  Google Scholar 

  3. Roscoe A, et al.: Anesth Analg. 2007, 104: 655-658. 10.1213/01.ane.0000255171.94527.c7

    Article  PubMed  Google Scholar 

Download references

Author information

Authors and Affiliations

Authors

Rights and permissions

This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.

Reprints and permissions

About this article

Cite this article

Végh, T., Enyedi, A., Takács, I. et al. Clinical experiences with a new endobronchial blocking device: the EZ-Blocker. Crit Care 15 (Suppl 1), P155 (2011). https://doi.org/10.1186/cc9575

Download citation

  • Published:

  • DOI: https://doi.org/10.1186/cc9575

Keywords