Evaluation of a continuous non-invasive arterial blood pressure monitoring device in comparison with an arterial blood pressure measurement in the ICU

Introduction

Due to a lower risk of complications, non-invasive monitoring methods gain importance. Measuring arterial blood pressure belongs to the standard hemodynamic monitoring. A newly developed continuous non-invasive arterial blood pressure (CNAP) measurement method is available and has been validated perioperatively [1]. We compared the CNAP monitoring device with invasive arterial blood pressure measurement (IBP) as the gold standard in critically ill patients.

Methods

We performed a prospective study on 49 critically ill patients at a medical ICU. All patients were sedated and mechanically ventilated (BIPAP, tidal volume 7 to 8 ml/kg ideal body weight). Furthermore, all patients were under vasopressor therapy. CNAP was applied on two fingers of the hand contralateral to the invasive arterial blood pressure catheter in the A. radialis. All measurements were digitally recorded with a sample frequency of 100 Hz, every pulse beat was automatically identified by an algorithm [2] and subsequently artefacts were removed from the datasets. The average recording time in each patient was 163 minutes (±37 minutes/patient).

Results

In total we analysed 500,000 beats. Overall we observed a bias in mean pressure of -7.49 mmHg with a standard deviation of 10.90 mmHg. The Bland-Altman plot (Figure 1) showed a uniform distribution of the variances over all measured blood pressure values and a good agreement of the mean blood pressure between CNAP and IBP. When analysing the data of each individual patient, larger differences were found. The bias ranged from 0.28 to 23.9 mmHg (median = -6.6 mmHg), with a standard deviation between 2.0 and 14.9 mmHg (median = 5.8 mmHg).

Figure 1
figure 1

Comparison between IBP and CNAP in 46 patients (50 beats per patient).

Conclusions

In our study we detected a good overall agreement between CNAP and IBP. The future perspective of this study is to investigate whether the continuous non-invasive blood pressure waveform is suitable for deriving further hemodynamic parameters of fluid responsiveness.

References

  1. Jeleazcov , et al.: Br J Anaesth. 2010, 105: 264-272. 10.1093/bja/aeq143

    Article  CAS  PubMed  Google Scholar 

  2. Zong , et al.: Comput Cardiol. 2003, 30: 259-262.

    Google Scholar 

Download references

Author information

Authors and Affiliations

Authors

Rights and permissions

This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.

Reprints and permissions

About this article

Cite this article

Smolle, K., Schmid, M. Evaluation of a continuous non-invasive arterial blood pressure monitoring device in comparison with an arterial blood pressure measurement in the ICU. Crit Care 15 (Suppl 1), P72 (2011). https://doi.org/10.1186/cc9492

Download citation

  • Published:

  • DOI: https://doi.org/10.1186/cc9492

Keywords