Evaluation of a continuous non-invasive arterial blood pressure monitoring device in comparison with an arterial blood pressure measurement in the ICU
Critical Care volume 15, Article number: P72 (2011)
Introduction
Due to a lower risk of complications, non-invasive monitoring methods gain importance. Measuring arterial blood pressure belongs to the standard hemodynamic monitoring. A newly developed continuous non-invasive arterial blood pressure (CNAP) measurement method is available and has been validated perioperatively [1]. We compared the CNAP monitoring device with invasive arterial blood pressure measurement (IBP) as the gold standard in critically ill patients.
Methods
We performed a prospective study on 49 critically ill patients at a medical ICU. All patients were sedated and mechanically ventilated (BIPAP, tidal volume 7 to 8 ml/kg ideal body weight). Furthermore, all patients were under vasopressor therapy. CNAP was applied on two fingers of the hand contralateral to the invasive arterial blood pressure catheter in the A. radialis. All measurements were digitally recorded with a sample frequency of 100 Hz, every pulse beat was automatically identified by an algorithm [2] and subsequently artefacts were removed from the datasets. The average recording time in each patient was 163 minutes (±37 minutes/patient).
Results
In total we analysed 500,000 beats. Overall we observed a bias in mean pressure of -7.49 mmHg with a standard deviation of 10.90 mmHg. The Bland-Altman plot (Figure 1) showed a uniform distribution of the variances over all measured blood pressure values and a good agreement of the mean blood pressure between CNAP and IBP. When analysing the data of each individual patient, larger differences were found. The bias ranged from 0.28 to 23.9 mmHg (median = -6.6 mmHg), with a standard deviation between 2.0 and 14.9 mmHg (median = 5.8 mmHg).
Conclusions
In our study we detected a good overall agreement between CNAP and IBP. The future perspective of this study is to investigate whether the continuous non-invasive blood pressure waveform is suitable for deriving further hemodynamic parameters of fluid responsiveness.
References
Jeleazcov , et al.: Br J Anaesth. 2010, 105: 264-272. 10.1093/bja/aeq143
Zong , et al.: Comput Cardiol. 2003, 30: 259-262.
Author information
Authors and Affiliations
Rights and permissions
This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.
About this article
Cite this article
Smolle, K., Schmid, M. Evaluation of a continuous non-invasive arterial blood pressure monitoring device in comparison with an arterial blood pressure measurement in the ICU. Crit Care 15 (Suppl 1), P72 (2011). https://doi.org/10.1186/cc9492
Published:
DOI: https://doi.org/10.1186/cc9492