Comparison of cardiac index: LiDCOrapid and PiCCOplus in the ICU

Introduction

This study aims to compare two arterial pressure wave-form monitors: the nomogram scaled LiDCOrapid (LiDCO Ltd, London, UK) with the calibrated PiCCOPlus, (Pulsion, Munich, Germany), to determine agreement for cardiac index (CI) measurement and trending during positional changes of passive leg raise test (PLRT) and volume expansion in the SICU.

Methods

We recruited 20 patients who had undergone major abdominal or neurosurgery and 10 patients in the SICU with progressive circulatory instability. The femoral artery was cannulated to obtain the arterial blood pressure waveform. Simultaneous measurements were made at four time points, M1 to M4: (M1) baseline, (M2) after PLRT, (M3) baseline (M4) after 500 ml Tetraspan® 6% over 10 minutes via pressure infusion. The PiCCO was calibrated via transpulmonary thermodilution at each time point.

Results

Data were collected from 30 patients, age 31 to 90, ASA 2 (2), ASA 3 (24) or ASA 4 (4), BSA 1.54 to 2.52 m2. CI ranged from 1.5 to 7.2 l/minute/m2 for PiCCO and from 1.5 to 7.1 l/min/m2 for LiDCO. Regression plots were made at each time point and show good agreement across the full range of CI values (r2 = 0.89 to 0.95). Bland-Altman analysis at each time point found low bias (10 to 50 ml/min/m2) and acceptable limits of agreement (16 to 30%), with the greatest difference occurring after the PLRT. Trending analysis was conducted by four-quadrant plot concordance assessment using an optimised exclusion zone of <5% ΔCI on changes at timepoints M2 to M4 relative to baseline (M1). Concordance was calculated as 97.8% overall agreement (44/45) for ΔCI >5%. Regression analysis found a high degree of correlation (r2 = 0.86 to 0.92) and all intercepts equal to 0.

Conclusions

In a heterogeneous patient population, LiDCOrapid CI values are in agreement with PiCCO CI values according to the accepted standard of ± 30% with minimal bias. Trending analysis showed excellent concordance of 97.8%, which meets the recently proposed standard of >90% [1]. The LiDCOrapid is a valid measure of CI and trends in CI. It is easier to set up, does not require central venous access, is independent of the arterial site and can be used both intraoperatively in the OR and in the ICU.

References

  1. Critchley LA, et al.: Anesth Analg. 2010, 111: 1180-1192. 10.1213/ANE.0b013e3181f08a5b

    Article  PubMed  Google Scholar 

Download references

Author information

Authors and Affiliations

Authors

Rights and permissions

This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.

Reprints and permissions

About this article

Cite this article

Brass, P., Mills, E., Latza, J. et al. Comparison of cardiac index: LiDCOrapid and PiCCOplus in the ICU. Crit Care 15 (Suppl 1), P62 (2011). https://doi.org/10.1186/cc9482

Download citation

  • Published:

  • DOI: https://doi.org/10.1186/cc9482

Keywords