Prediction of fluid responsiveness in septic shock patients: comparing automated pulse pressure variation by IntelliVue MP monitor and stroke volume variation by FloTrac™/Vigileo™

Introduction

The aim of this study was to assess and compare the ability of the automatically and continuously measured pulse pressure variation (PPV) obtained by an IntelliVue MP monitor and stroke volume variation (SVV) measured by FloTrac™/Vigileo™ to predict fluid responsiveness in septic shock patients.

Methods

We conducted a prospective study on 42 mechanically ventilated septic shock patients. SVV, PPV and other hemodynamic data were recorded before and after fluid administration with 500 ml of 6% tetrastarch. Responders were defined as patients with an increase in cardiac index ≥15% after fluid loading.

Results

The agreement (mean bias ± SD) between PPV and SVV was -0.59 ± 1.72% (Figure 1). The baseline PPV correlated with the baseline SVV (r = 0.96, P < 0.001). Twenty-seven (64.3%) patients were classified as fluid responders. PPV and SVV were significantly higher in responders than in nonresponders (16.2 ± 4.9 vs. 7.1 ± 2% and 15.3 ± 4.3 vs. 6.9 ± 1.9%, respectively, P < 0.001 for both). There was no difference between the area under the receiver operating characteristic curves of PPV (0.983) and SVV (0.99). The optimal threshold values to predicting fluid responsiveness were 10% for PPV (sensitivity 92.6%, specificity 86.7%) and 10% for SVV (sensitivity 92.6%, specificity 100%).

Figure 1
figure 1

Bland-Altman analysis for the agreement between SVV and PPV.

Conclusions

The automated PPV, obtained by the Intellivue MP monitor, and the SVV, obtained by FloTrac™/Vigileo™, showed comparable performance in terms of predicting fluid responsiveness in mechanically ventilated patients with septic shock.

References

  1. Cannesson M, et al.: Anesth Analg. 2008, 106: 1195-1200. 10.1213/01.ane.0000297291.01615.5c

    Article  PubMed  Google Scholar 

  2. Derichard A, et al.: Br J Anaesth. 2009, 103: 678-684. 10.1093/bja/aep267

    Article  CAS  PubMed  Google Scholar 

Download references

Author information

Authors and Affiliations

Authors

Rights and permissions

This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.

Reprints and permissions

About this article

Cite this article

Khwannimit, B., Bhurayanontachai, R. Prediction of fluid responsiveness in septic shock patients: comparing automated pulse pressure variation by IntelliVue MP monitor and stroke volume variation by FloTrac™/Vigileo™. Crit Care 15 (Suppl 1), P53 (2011). https://doi.org/10.1186/cc9473

Download citation

  • Published:

  • DOI: https://doi.org/10.1186/cc9473

Keywords