Using flow-rate limiting bag-valve-mask device without training
Critical Care volume 14, Article number: P305 (2010)
Introduction
The Smart Bag (SB, O-Two Medical Technologies Inc., Canada) is a bag-valve-mask resuscitator (BVM) equipped with a valve limiting flow-rate and peak airway pressure to decrease hyperventilation and gastric insufflation. We studied this device used by hospital staff without training to evaluate the possibility to change the standard BVM to the SB in hospital wards where positive pressure ventilation is rarely needed and training is sporadic.
Methods
The participants (12 physicians, 38 nurses) were randomised to use the standard BVM or SB to ventilate a resuscitation training manikin over 1 minute. The mechanism of the SB was shortly described in the SB group but no hands-on training was provided. Participants were asked to ventilate the patients as they would do if patients have a pulse but do not breathe. Tidal volumes were registered to a computer connected to the manikin.
Results
The medians of minute ventilation were 6.1 (interquartile range 2.6 to 8.1) and 3.9 (IQR 1.7 to 5.4) litres per minute in the SB and control groups, respectively (Figure 1). Hyperventilation >10 l/minute occurred only in the control group (P = 0.23).
Conclusions
Using the SB without previous hands-on training is possible for the majority of nursing and medical staff and decreases hyperventilation in comparison with the standard BVM.
Author information
Authors and Affiliations
Rights and permissions
About this article
Cite this article
Nurmi, J., Castrén, M. Using flow-rate limiting bag-valve-mask device without training. Crit Care 14 (Suppl 1), P305 (2010). https://doi.org/10.1186/cc8537