13C-stable isotope analyses technique for ICU patients
Critical Care volume 4, Article number: P133 (2000)
Full text
Introduction
By 13C isotope spectroscopy, a noninvasive diagnostic tool for specific functions of different organs is available [1]. A 13C labelled tracer is administered and is metabolised by a specific pathway to 13CO2. The increase of exhaled 13CO2 is analysed and gives an indication of a specific function of the organ under test. The availability of sensitive (resolution 0.3??) and stable isotope-selective infra-red (IR) spectrometers has replaced expensive mass-spectrometers. The present IR analysers rely on two conditions: 1) on co-operating patients who collect breath samples into containers and; 2) on equilibrated test gas samples.
Methods
Online monitoring of ventilated ICU patients over a long period requires an automated collection of the exhaled gases and an equilibration of the CO2 content without compromising the lung ventilator set-up. The sample should only contain exhaled gases, a dilution would increase the necessary sensitivity of the analyser. A collection and mixing device has been developed that collects the exhaled gases and does not interfere with the safety and monitoring concept of the lung ventilator. The unit collects the sample gas from an Evita ventilator (Dräger) and supplies it to an Infra-Red Isotope-selective Spectrometer (IRIS, Wagner) which has been modified to analyse a continuous gas stream. The expiratory concentrations of 13CO2 and 12CO2, the ratio of 13CO2 / 12CO2 as well as the cumulated hourly percentage of the recovery of the administered 13C label are determined by evaluation IRIS software (V2.3 1998, Wagner) and stored on a computer.
Results
The risk analysis and technical set-up have been reviewed and certified by the TÜV-Rheinland. The collection unit connecting the lung ventilator to the analyser has proven not to interfere with the functionality of the therapeutic device. The set-up was validated in volunteers and in ICU patients. Results of a first study determining the gastric emptying by isotope techniques will be discussed in parallel on this congress.
Conclusion
The stable IR isotope spectroscopy can be applied to ventilated patients. This allows a non-invasive focus e.g. on liver enzyme activities and on other organ specific problems.
References
Ghoos Y: 13 CO 2 breath tests at the laboratory `digestion-absorption'. University Hospital Gasthuisberg, Leuven 1996.
Author information
Authors and Affiliations
Rights and permissions
About this article
Cite this article
Hölscher, U., Gruchmann, T., Goeters, C. et al. 13C-stable isotope analyses technique for ICU patients. Crit Care 4 (Suppl 1), P133 (2000). https://doi.org/10.1186/cc853
Published:
DOI: https://doi.org/10.1186/cc853
Keywords
- Lung Ventilator
- Liver Enzyme Activity
- Evaluation Iris
- Noninvasive Diagnostic Tool
- Monitoring Concept