Non-invasive cardiac output monitoring in children: clinical validation
- Published: 1 March 2010
Keywords
- Cardiac Output
- Cardiomyopathy
- Congenital Heart Disease
- Dilate Cardiomyopathy
- Interventional Procedure
Introduction
Continuous noninvasive cardiac output (CO) provides valuable information for patient management. Non-invasive cardiac output monitoring (NICOM) measures CO based on chest bioreactance and is validated in adults [1]. Validated data in children are lacking. Our objective was to evaluate NICOM in children with pulmonary artery catheter thermodilution (PAC) as reference.
Methods
Paired CO values using NICOM and TD were recorded during cardiac catheterization in children with congenital heart disease. Children with intracardiac or extracardiac shunts were excluded. PAC was inserted through the femoral vein and CO was measured after bolus injection of 5 ml iced saline. NICOM was connected in accordance with the manufacturers' directrix.
Results
Bland-Altman analysis.
Conclusions
CO measurements with NICOM and PAC show a significant correlation. Bland-Altman demonstrates an acceptable agreement; however, the limits of agreement are broad. Depending on the CO range, NICOM reveals a trend to systematically overestimate or underestimate CO. Additional studies in larger and more heterogeneous pediatric patient populations are warranted for further validation.
Authors’ Affiliations
References
- Squara P, et al: Intensive Care Med. 2007, 33: 1191-1194. 10.1007/s00134-007-0640-0.View ArticleGoogle Scholar