Changes in regional distribution of lung sounds as a function of positive end-expiratory pressure

Abstract

Introduction

Automated mapping of lung sound distribution is a novel area of interest currently investigated in mechanically ventilated, critically ill patients. The objective of the present study was to assess changes in thoracic sound distribution resulting from changes in positive end-expiratory pressure (PEEP). Repeatability of automated lung sound measurements was also evaluated.

Methods

Regional lung sound distribution was assessed in 35 mechanically ventilated patients in the intensive care unit (ICU). A total of 201 vibration response imaging (VRI) measurements were collected at different levels of PEEP between 0 and 15 cmH2O. Findings were correlated with tidal volume, oxygen saturation, airway resistance, and dynamic compliance. Eighty-two duplicated readings were performed to evaluate the repeatability of the measurement.

Results

A significant shift in sound distribution from the apical to the diaphragmatic lung areas was recorded when increasing PEEP (paired t-tests, P < 0.05). In patients with unilateral lung pathology, this shift was significant in the diseased lung, but not as pronounced in the other lung. No significant difference in lung sound distribution was encountered based on level of ventilator support needed. Decreased lung sound distribution in the base was correlated with lower dynamic compliance. No significant difference was encountered between repeated measurements.

Conclusions

Lung sounds shift towards the diaphragmatic lung areas when PEEP increases. Lung sound measurements are highly repeatable in mechanically ventilated patients with various lung pathologies. Further studies are needed in order to fully appreciate the contribution of PEEP increase to diaphragmatic sound redistribution.

Introduction

The use of acoustic monitoring technology offers the potential for a radiation-free, noninvasive bedside assessment of lung abnormality in patients during their stay in the intensive care unit (ICU). Correlation between breath sound recordings and regional distribution of pulmonary ventilation has been previously established, particularly in studies conducted by Ploy-Song-Sang and colleagues and other groups who compared acoustic findings with data obtained with radioactive gases [1–3]. The effect of airflow and volume on the amplitude and spectral content of breath sounds has been extensively studied in healthy [4–9] and diseased lungs [10–12]. Furthermore, several studies assessed the effect of changes of mechanical ventilation on lung sound distribution in animal models [13–17]. Räsenen and colleagues reported that the acoustic changes associated with oleic acid-induced lung injury allow monitoring of its severity and distribution [13] and that acute lung injury (ALI) causes regional acoustic transmission abnormalities that are reversed during alveolar recruitment with positive end-expiratory pressure (PEEP) [14]. Recently, Vena and colleagues reported a reduction of amplitude and a change in spectral characteristics of normal lung sounds when increasing PEEP in mechanically ventilated pigs [15]. Finally, recording of crackle-sound during mechanical ventilation was employed to monitor lung recruitment–derecruitment in a porcine model [16, 17].

The experience on acoustic monitoring in mechanically ventilated patients is limited [18, 19] and only preliminary investigations were conducted to assess changes in regional distribution of lung sound as a function of changes in mechanical ventilator setting [20, 21]. Waitman and colleagues classified breath sounds recorded in an intensive care setting using different neural network configurations [22], and a computerized respiratory sound monitor was used to detect wheezes in pediatric ICU [23]. Detection of endobronchial [24–26] and esophageal [27] intubation using lung sound monitoring during anesthesia was also described. Dellinger and colleagues recently reported the use of an acoustic-based imaging device to map the geographical distribution of breath sound as a function of mechanical ventilation mode [28]. Changes in lung sound distribution map during recruitment maneuver and PEEP increase were also reported in four abstracts [29–32]. These findings suggest that breath sound information can be used to evaluate lung condition during mechanical ventilation; however, information regarding lung sound monitoring to adjusted PEEP levels is lacking. PEEP setting is widely used by physicians and respiratory therapists in order to improve gas exchange, mainly in patients with severe hypoxic respiratory failure such as acute respiratory distress syndrome (ARDS) and ALI [33] while preventing end-expiratory alveolar collapse [34] and inspiratory overinflation [35]. In practice PEEP setting is adjusted to patient condition up to several times a day, although no standardized method to adjust PEEP has been accepted to date. The first step in the evaluation of a new approach is to assess if a change in PEEP induces any change in the measurement.

The aim of the present study was to evaluate the effect of changes in PEEP on the regional distribution of lung sounds as recorded by vibration response imaging (VRI), an acoustic monitoring technology that creates a dynamic two-dimensional functional image of lung sound distribution during mechanical ventilation. Repeatability of lung sound measurements was also evaluated.

Materials and methods

Patients

The study was performed in the general ICU of the Rabin Medical Center in Petach-Tikva, Israel. The study protocol was approved by the Institutional Review Board and informed consent was obtained from all patients or their next-of-kin. Intra-individual differences in lung sound measurements were investigated at different levels of PEEP in a prospective trial.

Inclusion and exclusion criteria

Patients enrolled in the study were adults (18 to 85 years old) with a body mass index greater than 21. Exclusion criteria included a body habitus or skin condition that would interfere with sensor placement; the presence of a cardiac pacemaker, implantable defibrillator or artificial heart valve; or pregnancy.

Study design

The levels of PEEP and the fraction of inspired oxygen (FiO2) were adjusted according to clinical requirements. Among the 35 patients enrolled in this study, one was recorded at PEEP 5 and 10 cmH2O and 34 at PEEP 0, 5 and 10 cmH2O. Fifteen of these 34 patients were also recorded at PEEP 15 cmH2O. In 28 patients, PEEP was assigned from low to high level. In order to assess any effect due to the lack of randomization, PEEP levels were applied in a random order in a subgroup of patients (n = 7). At the later stage of the protocol, repeatability was tested on 26 patients for whom two repeated consecutive measurements were performed at the same level of PEEP under the same conditions, over a period of time not exceeding five minutes. No recording was excluded from the repeatability study. Measurements at different PEEP levels were performed at an interval of at least five minutes. No intervention, except for changes in PEEP, was allowed by the protocol. Mode of mechanical ventilation, tidal volume (VT), respiratory rate (RR), partial arterial pressure of oxygen (PaO2), FiO2, oxygen saturation (SpO2), airway resistance (Raw) and online dynamic compliance (Cdyn) as provided by the ventilator were documented. Three consecutive measurements of Cdyn were averaged in order to reduce variability. To keep consistency and ensure that timing between spontaneous and controlled cycles do not affect the results, the spontaneous breath was used whenever available (31 out of 35 patients, 89%), including in synchronized intermittent mandatory ventilation mode (SIMV).

Recording procedure

A schema of the apparatus is provided in Figure 1. The recordings were performed using a VRIxv™ device (Deep Breeze Ltd., Or-Akiva, Israel) with two arrays of six rows by three columns sensors or microphones similar to those used in digital stethoscopes. The recordings were made in supine position with a bed angle between 30 to 45°. The arrays were positioned posterior to the patient's back using a disposable positioning unit to reduce risk of cross-contamination. Morphological landmarks such as spine and scapula were used in order to ensure accurate and repeatable placement of the sensor arrays. Excessive secretions were removed by endotracheal and oral suctioning before each series of recordings. Airway pressure and flow waveforms were sampled from the ventilator using a proximal flow sensor inserted in the patient's circuit. As displayed in Figure 2, these waveforms were synchronized with the sound energy graph representing the average sound energy in both lungs. Each recording lasted for 20 seconds of acquisition time, followed by 40 seconds of processing time, and was stored digitally on the device for later review and analysis.

Figure 1
figure 1

Schematic diagram describing the elements of the system. The patient lies on the acoustic sensor array and the flow sensor is inserted in the breathing circuit. The vibration response imaging (VRI) system collects acoustic information simultaneously from the sensor array and pressure and flow waveforms from the ventilator.

Figure 2
figure 2

An example of acoustic data as displayed for a recording obtained from a 77-year-old male with myasthenia gravis. A representative peak-inspiratory image (left panel); synchronized sound energy graph and ventilator airway pressure and flow waveforms (middle panel); sound energy distribution in the six lung regions as automatically provided by the software in percentage of weighted pixel count (right panel). VRI = vibration response imaging.

Measurement output

The output of the measurement consisted of a normalized dynamic image synchronized with airway pressure and flow waveforms, revealing the geographical distribution of sound energy during the respiratory cycle. Because of image normalization, the measurement output aimed at describing the relative airflow distribution in the lung rather than the absolute volume. As described in detail by Dellinger and colleagues [36], this dynamic image was created from a series of gray-scale still images or frames with each frame representing 0.17 seconds of sound energy. The digitized acoustic signals were band-pass filtered between 150 and 250 Hz to remove heart and muscle sounds; median filtering was applied to suppress impulse noise, and truncation of samples above an automatically determined signal-to-noise threshold was performed. Sound energy was obtained following down-sampling.

Recording quality was assessed according to pre-determined criteria [28]. The graph representing the average sound energy as a function of time throughout the respiratory cycle in both lungs was displayed underneath the dynamic image. Each 20 second measurement included up to 10 respiratory cycles. A normalized representative frame (or map) at peak-inspiratory flow was automatically selected and displayed on the screen (Figure 2). This map was also quantified by the software and presented as the percentage of weighted pixels in six lung regions: upper right (UR); middle right (MR); lower right (LR); upper left (UL); middle left (ML) and lower left (LL), up to a total of 100%. According to the recording procedure, 'lower' corresponded to the diaphragmatic lung region and 'upper' to the apical lung region. The percentage of sound energy in the total right lung (TR = UR + MR + LR) and total left lung (TL = UL + ML + LL) were also documented. The apico-diaphragmatic ratio (ADR), defined as the ratio between the lung sound distribution in the apical lung areas (UR + UL) and the diaphragmatic lung areas (LR + LL) (ADR = (UL + UR)/(LL + LR)), was used to assess changes in lung sound distribution along the cephalocaudal axis. A larger ratio suggested increased sound in the apical areas and a smaller ratio increased sound in the diaphragmatic areas. Distribution was considered more heterogeneous if the difference was larger than two; this threshold was derived from experience with healthy patients in supine positions.

Statistical analysis

Data are presented as the mean ± standard deviation. When two recordings were performed under identical conditions, the second recording was used for all analysis not related to the repeatability aspect of the study. For samples including more than 30 measurements, paired student's t-test (Microsoft® Office Excel 2003, Microsoft Corporation, Redmond, WA, USA) was used. For the analysis of samples including less than 30 measurements, Wilcoxon matched-pairs signed-ranks test, Friedman two-way analysis on ranks test and Wilcoxon two sample test were used (IFA Services Statistics, Amsterdam, Holland). Friedman test was used to compare three or more paired groups. Coefficients of determination (R2) and coefficients of variation (CV) were used to test repeatability. A P < 0.05 was considered significant.

Results

A total of 35 mechanically ventilated patients (26 males, 9 females, age 62 ± 20 years) were enrolled in the study between April 2007 and January 2008. Patients were ventilated using one of two types of ventilators (Puritan Bennett, Tyco Healthcare, Mansfield, MA, USA; Evita XL or Evita 4, Draeger, Lübeck, Germany). The majority of the patients (n = 26; 74%) were mechanically ventilated on pressure support ventilation mode (PSV) with a level between 8 and 24 cmH2O (mean 14 ± 4 cmH2O). Six patients (17%) were ventilated using SIMV. The rest of the patients (n = 3; 8%) were ventilated with other modes of mechanical ventilation. Patients were not deeply sedated and none were paralyzed.

A total of 201 valid recordings were performed on the 35 patients. No adverse event related to the measurement was registered. Ten recordings (less than 5% of the overall data) were excluded from the analysis based on pre-determined criteria as mentioned above [28]. Poor recording quality was confirmed by an average sound energy level below a pre-defined threshold (< 1 in the energy bar of the imaging display). Reasons for mechanical ventilation of these 35 patients are described in Table 1. Chest radiography results revealed that 19 of these patients had bilateral disease, 13 had unilateral lung pathology inducing decreased lung sounds (i.e. one-lung atelectasis, pneumothorax, or pleural fluid) and three had normal lungs. Average VT was 551 ± 126 mL, SpO2 97 ± 3%, RR 21 ± 7 breaths/minute, Cdyn 60 ± 42 mL/mbar, and Raw 16 ± 5 mbar L/second. These parameters did not significantly change with PEEP (Friedman test, paired groups).

Table 1 Reason for mechanical ventilation in 35 patients

Paired analysis conducted on the 34 patients for which recordings at PEEP 0, 5, and 10 cmH2O were available revealed that the proportion of sound energy in the diaphragmatic lung regions (LR and LL) was significantly increased with PEEP (P < 0.05, paired t-test), while the proportion of sound energy in the apical lung regions (UR and UL) was decreased (P < 0.05 in UL, paired t-test). The proportion of energy in the middle areas of the lungs (MR and ML) did not significantly change with PEEP (Figure 3). No additional shift was detected at PEEP 15 cmH2O (n = 15, Wilcoxon matched paired test). In patients with unilateral lung pathology (n = 13), the increase in sound energy in the diaphragmatic lung regions was significant in the diseased lung (7 ± 6% at PEEP 0 cmH2O versus 10 ± 7% at PEEP 10 cmH2O, P = 0.01, Wilcoxon matched-pairs) but not significant in the other lung (14 ± 8% at PEEP 0 cmH2O versus 15 ± 9% at PEEP 10 cmH2O, Wilcoxon matched-pairs). In patients with bilateral lung pathology (n = 21), the increase was significant in both lungs (P = 0.04).

Figure 3
figure 3

Mean ± standard deviation of sound energy distribution in 34 mechanically-ventilated patients recorded at three levels of PEEP (0, 5 and 10 cmH2O). Significant P values are indicated (paired t-tests). LL = lower left; LR = lower right; ML = middle left; MR = middle right; PEEP = positive end-expiratory pressure; UL = upper left; UR = upper right.

The majority of the patients were ventilated on PSV or SIMV, spontaneous diaphragmatic activity was present in most of the patients. In order to assess the extent of this confounding factor, analysis was conducted according to the level of ventilator support provided to the patients. Patients were divided into two subsets according to the level of ventilator support needed (PSV < 15 cmH2O and PSV > 15 cmH2O). Sound energy distribution was compared between the two groups at each level of PEEP. No significant difference was detected (Wilcoxon two sample tests).

As shown in Figure 4, per patient analysis revealed that when increasing PEEP from 0 to 10 cmH2O, sound energy distribution increased in the diaphragmatic lung areas in 76% of the patients (26 of 34). In these cases, a larger peak-inspiratory flow image was obtained at higher PEEP (examples in Figure 5a and 5b). In several patients, an asymmetrical change of lung sound energy distribution was recorded at PEEP 15 cmH2O (Figure 5c, d, and 5e). Comparisons between VT, SpO2, Cdyn and Raw at two different levels of ADR are summarized in Table 2. When adjusted for RR, no difference in VT, Raw and SpO2was encountered between the two levels of ADR. At RRs lower than 20 breaths/minute, Cdyn tended to be higher for recordings with increased energy in the lower lung regions (ADR < 2). This difference approached significance (P = 0.058).

Figure 4
figure 4

Individual sound energy distribution in diaphragmatic lung areas in 34 mechanically-ventilated patients recorded at PEEP levels 0 and 10 cmH2O. Sound energy distribution increased from 17 ± 11% to 23 ± 12% (P < 0.0001) in (a) 26 'responder' patients and decreased from 30 ± 17% to 27 ± 17% (P < 0.001) in (b) eight 'non-responder' patients. PEEP = positive end-expiratory pressure.

Figure 5
figure 5

Representative frames (or maps) at peak-inspiratory flow obtained from five individual patients at PEEP levels 0, 5, 10 and 15 cmH2O. (a) A 74-year-old female with respiratory failure. (b) A 19-year-old male with right pneumothorax. (c) A 83-year-old male with sternal wound infection. (d) A 77-year-old male with myasthenia gravis. (e) A 57-year-old male with acute pancreatitis. PEEP = positive end-expiratory pressure.

Table 2 Comparison between tidal volume, oxygen saturation, dynamic compliance, and airway resistance at two different levels of apico-diaphragmatic ratio

The repeatability of the measurement was assessed in 82 sets of double recordings obtained from 26 patients (20 double recordings at PEEP 0 cmH2O; 25 at PEEP 5 cmH2O; 26 at PEEP 10 cmH2O, and 11 at PEEP 15 cmH2O to a total of 164 recordings). Repeatability was performed by comparing the distribution of sound energy in each of the six lung regions of two repeated measurements, as well as in total left and right lungs. No significant difference was encountered between repeated measurements (paired t-test). Mean R2 obtained for the different lung regions was 0.93 ± 0.02 (range 0.91 to 0.95) with a CV equal to 1.7%.

Discussion

In this study, we used an acoustic-based monitoring system in order to assess possible shift in thoracic sound distribution during PEEP changes and to evaluate the repeatability of lung sound measurements in mechanically ventilated patients. Our results revealed a significant increase in sound distribution from the apical to the diaphragmatic lung areas when increasing PEEP from 0 to 10 cmH2O. This shift was especially pronounced in patients with severe lung pathology but was not affected by the level of pressure support needed. These statistical results were further supported by the analysis of the effect of PEEP on lung sound distribution in individual patients. As revealed in Figure 4, lung sound increased in the diaphragmatic lung areas in 76% of the patients.

The explanation for this acoustic phenomenon might be related to an increase in ventilation distribution in the diaphragmatic part of the lungs at higher levels of PEEP or to the effect of other PEEP-related physiologic factors, such as translocation of fluid from alveolar to interstitial spaces. A similar shift of lung sound distribution towards the base was recently described by Dellinger and colleagues [28], while changing mode of mechanical ventilation from volume control to pressure control and pressure support. The authors speculated that this shift was produced by a diaphragm-generated negative intrapleural pressure in pressure-targeted modes. The authors also proposed that the initial higher flow in pressure-targeted modes may serve to prime the proximal airway, allowing more time for slower, more laminar flow to produce a more homogenous distribution of air to lower lung regions. Correlation between lung condition and heterogeneity of lung sound distribution has been described in several additional studies. Bentur and colleagues [37] identified greater heterogeneity of lung sound distribution in pediatric patients with confirmed cases of foreign body aspiration when compared with healthy subjects. Lung sound heterogeneity was also described by Jean and colleagues [38] when comparing measurements performed on patients with normal lungs versus one diseased lung on chest radiograph. In agreement with the image normalization method used to generate the lung acoustic map, the authors explained that a larger image was considered to indicate a more homogeneous sound distribution and a smaller image a more focal distribution. In our study, the shift of lung sound distribution towards the base at PEEP 10 cmH2O was accompanied by an increase in the size of the peak-inspiratory flow image, in line with increased homogeneity of lung sound distribution. At PEEP level of 15 cmH2O, however, the lack of increased shift towards the base was adjunctive with a decrease in the size of the image as exemplified in Figures 5c to 5e. In light of the effect of PEEP elevation in lung sound distribution, comparison between measurements should be preferably performed when similar PEEP levels are applied.

Repeatability of the lung acoustic measurements was comparable with that reported in healthy subjects [8, 39, 40]. This result in patients mechanically ventilated in pressure support mode may be a priori unexpected, especially when considering the variability of VT anticipated in this mode. However, this finding confirms that normalization of the acoustic distribution map reduces the mere effect of changes in ventilator settings when these changes do not affect the relative airflow distribution. Figure 6a, representing normalized images recorded from the same patient ventilated with two different VT, and Figure 6b, representing normalized images recorded from the same patient ventilated with two different airflow rates, further illustrate this finding.

Figure 6
figure 6

Representative frames (or maps) at peak-inspiratory flow obtained from two patients ventilated with different ventilator settings. (a) A 72-year-old female with chronic obstructive pulmonary disease recorded at positive end-expiratory pressure (PEEP) level of 5 cmH2O and at two levels of tidal volume (VT; left = 330 mL, right = 560 mL); (b) A 24-year-old male with bilateral chest contusion recorded at PEEP level of 7 cmH2O, VT of 600 mL and at two levels of respiratory rate (RR) and inspiratory/expiratory ratio (i:e; left: i:e = 2:3 and RR = 12 breaths/minute, right: i:e = 5:2 and RR = 16 breaths/minute). TL = total left lung; TR = total right lung.

The scope of this study was limited because of a restrictive protocol. Enrollment of deeply sedated patients mechanically ventilated in volume-controlled mode of mechanical ventilation may have allowed controlling for VT and inspiratory flow but this was out of the scope of this protocol. Another protocol-related limitation included the effect of volume history which may interfere with VT distribution. Moreover, differentiation between diaphragmatic redistribution induced by PEEP and VT-induced recruitment may be difficult. The population was heterogeneous and further studies should be performed on a more homogeneous population allocated to specific lung disease categories with emphasis on ALI or ARDS. Moreover, the heterogeneity of the clinical conditions exhibited by the patients at the time of investigation may be a limitation of the present study. Despite normalization, airflow velocity of the ventilators may have affected the results and, considering its impact on VT distribution and dynamic hyperinflation, it would have been interesting to consider the peak flow values. Furthermore, the protocol did not include a comparison of lung sound distribution with more appropriate tools, such as computerized tomography, functional residual capacity, or electrical impedance tomography. This should be investigated in the future. Although sometimes useful in research and clinical practice [41, 42], the reliability of Cdyn is debatable, especially in non-paralyzed patients with non-uniform volume histories. Despite the fact that during pressure support mode of mechanical ventilation, Cdyn is particularly difficult to interpret, this parameter was used in this study because it was readily accessible in the scope of the protocol. In order to improve the accuracy of the measurement, three values were averaged at each time point. Finally, sound filtering to a band-pass of 150 to 250 Hz may have reduced the information as lung sound characteristics are contained in other frequency bands, especially above 250 Hz. Notwithstanding these limitations, the development of adjunctive technologies that assist in assessment of clinical benefits of PEEP and recruitment maneuver is still highly desirable [43, 44].

Conclusions

A shift in lung sound distribution from the apical to the diaphragmatic lung areas was observed during PEEP increase. This shift was not correlated with significant change in VT but was associated with an increase in Cdyn. High repeatability was obtained in this population. Further studies are needed in order to elucidate the mechanism of sound shift in relation to PEEP increment and to fully appreciate the contribution of PEEP increase to diaphragmatic sound redistribution.

Key messages

  • Sound distribution shifts from the apical to the diaphragmatic lung areas when increasing level of PEEP from 0 to 10 cmH2O.

  • Acoustic shift is not correlated with significant change in VT but is associated with increased Cdyn.

  • Sound distribution measurements are highly repeatable in mechanically ventilated patients.

Abbreviations

ADR:

apico-diaphragmatic ratio

ALI:

acute lung injury

ARDS:

acute respiratory distress syndrome

Cdyn:

dynamic compliance

CV:

coefficients of variation

FiO2:

fraction of inspired oxygen

ICU:

intensive care unit

LL:

lower left

LR:

lower right

ML:

middle left

MR:

middle right

PaO2:

partial arterial pressure of oxygen

PEEP:

positive end-expiratory pressure

PSV:

pressure support ventilation mode

R2:

coefficients of determination

Raw:

airway resistance

RR:

respiratory rate

SpO2:

oxygen saturation

SIMV:

synchronized intermittent mandatory ventilation

TL:

total left lung

TR:

total right lung

UL:

upper left

UR:

upper right

VRI:

vibration response imaging

VT:

tidal volume.

References

  1. Leblanc P, Macklem PT, Ross WR: Breath sounds and distribution of pulmonary ventilation. Am Rev Respir Dis 1970, 102: 10-16.

    CAS  PubMed  Google Scholar 

  2. Ploy-Song-Sang Y, Martin RR, Ross WR, Loudon RG, Macklem PT: Breath sounds and regional ventilation. Am Rev Respir Dis 1977, 116: 187-199.

    CAS  PubMed  Google Scholar 

  3. Ploy-Song-Sang Y, Macklem PT, Ross WR: Distribution of regional ventilation measured by breath sounds. Am Rev Respir Dis 1978, 117: 657-664.

    CAS  PubMed  Google Scholar 

  4. Kraman SS: The relationship between airflow and lung sound amplitude in normal subjects. Chest 1984, 86: 225-229. 10.1378/chest.86.2.225

    Article  CAS  PubMed  Google Scholar 

  5. Gavriely N, Cugell DW: Airflow effects on amplitude and spectral content of normal breath sounds. J Appl Physiol 1996, 80: 5-13.

    CAS  PubMed  Google Scholar 

  6. Kiyokawa H, Pasterkamp H: Volume-dependent variations of regional lung sound, amplitude, and phase. J Appl Physiol 2002, 93: 1030-1038.

    Article  PubMed  Google Scholar 

  7. Bergstresser T, Ofengeim D, Vyshedskiy A, Shane J, Murphy R: Sound transmission in the lung as a function of lung volume. J Appl Physiol 2002, 93: 667-674.

    Article  CAS  PubMed  Google Scholar 

  8. Maher TM, Gat M, Allen D, Devaraj A, Wells AU, Geddes DM: Reproducibility of dynamically represented acoustic lung images from healthy individuals. Thorax 2008, 63: 542-548. 10.1136/thx.2007.086405

    Article  PubMed Central  CAS  PubMed  Google Scholar 

  9. Yigla M, Gat M, Meyer JJ, Friedman PJ, Maher TM, Madison JM: Vibration response imaging technology in healthy subjects. AJR Am J Roentgenol 2008, 191: 845-852. 10.2214/AJR.07.3151

    Article  PubMed  Google Scholar 

  10. Murphy RL, Vyshedskiy A, Power-Charnitsky VA, Bana DS, Marinelli PM, Wong-Tse A, Paciej R: Automated lung sound analysis in patients with pneumonia. Respir Care 2004, 49: 1490-1497.

    PubMed  Google Scholar 

  11. Mor R, Kushnir I, Meyer JJ, Ekstein J, Ben-Dov I: Breath sound distribution images of patients with pneumonia and pleural effusion. Respir Care 2007, 52: 1753-1760.

    PubMed  Google Scholar 

  12. Kramer MR, Raviv Y, Hardoff R, Shteinmatz A, Amital A, Shitrit D: Regional breath sound distribution analysis in single-lung transplant recipients. J Heart Lung Transplant 2007, 26: 1149-1154. 10.1016/j.healun.2007.07.039

    Article  PubMed  Google Scholar 

  13. Räsenen J, Gavriely N: Detection of porcine oleic acid-induced acute lung injury using pulmonary acoustics. J Appl Physiol 2002, 93: 51-57.

    Article  Google Scholar 

  14. Räsenen J, Gavriely N: Response of acoustic transmission to positive airway pressure therapy in experimental lung injury. Intensive Care Med 2005, 31: 1434-1441. 10.1007/s00134-005-2745-7

    Article  Google Scholar 

  15. Vena A, Perchiazzi , Giuliani R, Fiore T, Hedenstierna G: Acoustic effects of positive end-expiratory pressure on normal lung sounds in mechanically ventilated pigs. Clin Physiol Funct Imaging 2006, 26: 45-53. 10.1111/j.1475-097X.2005.00652.x

    Article  PubMed  Google Scholar 

  16. Peták F, Habre W, Babik B, Tolnai J, Hantos Z: Crackle-sound recording to monitor airway closure and recruitment in ventilated pigs. Eur Respir J 2006, 27: 808-816. 10.1183/09031936.06.00105005

    Article  PubMed  Google Scholar 

  17. Vena A, Perchiazzi G, Rylander C, Giuliani R, Fiore T, Magnusson A, Hedenstierna : Breath sound analysis detects injury and recruitment in the lung during mechanical ventilation. Intensive Care Med 2008,34(Suppl 1):549.

    Google Scholar 

  18. Hubmayr RD: The times are a-changin' should we hang up the stethoscope? Anesthesiology 2004, 100: 1-2. 10.1097/00000542-200401000-00002

    Article  PubMed  Google Scholar 

  19. Lichtenstein D, Goldstein I, Mourgeon E, Cluzel P, Grenier P, Rouby J-J: Comparative diagnostic performances of auscultation, chest radiography, and lung ultrasonography in acute respiratory distress syndrome. Anesthesiology 2004, 100: 9-15. 10.1097/00000542-200401000-00006

    Article  PubMed  Google Scholar 

  20. Cinel I, Jean S, Dellinger RP: Dynamic lung imaging techniques in mechanically ventilated patients. In Yearbook of Intensive Care and Emergency Medicine. Edited by: Vincent JL. Heidelberg: Springer-Verlag; 2007:373-380.

    Google Scholar 

  21. Lev S, Singer P, Glickman YA: Vibration response imaging: a novel technology for lung monitoring in critically ill patients. In Yearbook of Respiratory Care Clinics and Applied Technologies. Edited by: Esquinas A. Murcia: World Federation of Respiratory Care and Applied Technologies; 2008:530-539.

    Google Scholar 

  22. Waitman LR, Clarkson KP, Barwise JA, King PH: Representation and classification of breath sounds recorded in an intensive care setting using neural networks. J Clin Monit Comput 2000, 16: 95-105. 10.1023/A:1009934112185

    Article  CAS  PubMed  Google Scholar 

  23. Prodhan P, Dela Rosa RS, Shubina M, Haver KE, Matthews BD, Buck S, Kacmarek RM, Noviski NN: Wheeze detection in the pediatric intensive care unit: comparison among physician, nurses, respiratory therapists, and a computerized respiratory sound monitor. Respir Care 2008, 53: 1304-1309.

    PubMed  Google Scholar 

  24. Tejman-Yarden S, Lederman D, Eilig I, Zlotnik A, Weksler N, Cohen A, Gurman GM: Acoustic monitoring of double-lumen ventilated lungs for the detection of selective unilateral lung ventilation. Anesth Analg 2006, 103: 1489-1493. 10.1213/01.ane.0000240909.48774.49

    Article  PubMed  Google Scholar 

  25. Tejman-Yarden S, Zlotnik A, Weizman L, Tabrikian J, Cohen A, Weksler N, Gurman GM: Acoustic monitoring of lung sounds for the detection of one-lung intubation. Anesth Analg 2007, 105: 397-404. 10.1213/01.ane.0000267263.40505.c0

    Article  CAS  PubMed  Google Scholar 

  26. Jean S, Cinel I, Gratz I, Tay C, Lotano V, Deal E, Parillo JE, Dellinger RP: Image-based monitoring of one-lung ventilation. Eur J Anaesthesiol 2008, 20: 1-7.

    Google Scholar 

  27. Cinel I, Jean S, Tay C, Gratz I, Deal E, Parillo JE, Dellinger RP: Case report: vibration response imaging findings following inadvertent esophageal intubation. Can J Anaesth 2008, 55: 172-176. 10.1007/BF03016092

    Article  PubMed  Google Scholar 

  28. Dellinger RP, Jean S, Cinel I, Tay C, Rajanala S, Glickman YA, Parillo JE: Regional distribution of acoustic-based lung vibration as a function of mechanical ventilation mode. Critical Care 2007, 11: R26. 10.1186/cc5706

    Article  PubMed Central  PubMed  Google Scholar 

  29. Trivedi H, Cinel I, Jean S, Tay C, Durflinger P, Parillo JE, Dellinger RP: The role of vibration response imaging in the titration of PEEP in a mechanically ventilated patient with acute respiratory distress syndrome. Crit Care Med 2007,34(Suppl 12):608.

    Google Scholar 

  30. Cinel I, Dellinger RP, Jean S, Glickman YA, Parillo JE: Assessment of the effectiveness of lung recruitment and PEEP setting by vibration response imaging. Crit Care 2006,10(Suppl 1):15. 10.1186/cc4362

    Article  Google Scholar 

  31. Lev S, Kagan I, Grinev M, Cohen J, Singer P: Positive end-expiratory pressure-induced changes of the vibration response image. Crit Care 2008,12(Suppl 2):299.

    Article  Google Scholar 

  32. Lev S, Cohen J, Kagan I, Grinev M, Singer P: Lung sound distribution shifts to the lower lung regions with increased PEEP. Intensive Care Med 2008,34(Suppl 1):545.

    Google Scholar 

  33. Talmor D, Sarge T, Malhotra A, O'Donnell CR, Ritz R, Lisbon A, Novack V, Loring SH: Mechanical ventilation guided by esophageal pressure in acute lung injury. N Engl J Med 2008, 359: 2095-104. 10.1056/NEJMoa0708638

    Article  PubMed Central  CAS  PubMed  Google Scholar 

  34. Pelosi P, Caironi P, Bottino N, Gattinoni L: Positive end expiratory pressure in anesthesia. Minerva Anestesiol 2000, 66: 297-306.

    CAS  PubMed  Google Scholar 

  35. Nieszkowska A, Lu Q, Vieira S, Elman M, Fetita C, Rouby JJ: Incidence and regional distribution of lung overinflation during mechanical ventilation with positive end-expiratory pressure. Crit Care Med 2004, 32: 1496-1503. 10.1097/01.CCM.0000130170.88512.07

    Article  PubMed  Google Scholar 

  36. Dellinger PR, Parrillo JE, Kushnir A, Rossi M, Kushnir I: Dynamic visualization of lung sounds with a vibration response device: a case series. Respiration 2008, 75: 60-72. 10.1159/000103558

    Article  PubMed  Google Scholar 

  37. Bentur L, Livnat G, Husein D, Pollack S, Rotschild M: Dynamic visualization of breath sounds distribution in suspected foreign body aspiration. J Bronchology 2007, 14: 156-161. 10.1097/LBR.0b013e3180959ad8

    Article  Google Scholar 

  38. Jean S, Cinel I, Tay C, Parrillo JE, Dellinger RP: Assessment of asymmetric lung disease in intensive care unit patients using vibration response imaging. Anesth Analg 2008, 107: 1243-1247. 10.1213/ane.0b013e3181804a99

    Article  PubMed  Google Scholar 

  39. Mahagnah M, Gavriely N: Repeatability of measurements of normal lung sounds. Am J Respir Crit Care Med 1994, 149: 477-481.

    Article  CAS  PubMed  Google Scholar 

  40. Sánchez I, Vizcaya C: Tracheal and lung sounds repeatability in normal adults. Respir Med 2003, 97: 1257-1260. 10.1016/S0954-6111(03)00251-8

    Article  PubMed  Google Scholar 

  41. Suarez-Sipmann F, Böhm SH, Tusman G, Pesch T, Thamm O, Reissmann H, Reske A, Magnusson A, Hedenstierna G: Use of dynamic compliance for open lung positive end-expiratory pressure titration in an experimental study. Crit Care Med 2007, 35: 214-221. 10.1097/01.CCM.0000251131.40301.E2

    Article  PubMed  Google Scholar 

  42. Suarez-Sipmann F: Titrating open lung PEEP in acute lung injury. A clinical method based on changes in dynamic compliance. In Digital Comprehensive Summaries of Uppsala Dissertations from the Faculty of Medicine 313. Acta Universitatis Upsaliensis, Uppsala; 2008.

    Google Scholar 

  43. Mercat A, Richard JC, Vielle B, Jaber S, Osman D, Diehl JL, Lefrant JY, Prat G, Richecoeur J, Nieszkowska A, Gervais C, Baudot J, Bouadma L, Brochard L, Expiratory Pressure (Express) Study Group: Positive end-expiratory pressure setting in adults with acute lung injury and acute respiratory distress syndrome: a randomized controlled trial. JAMA 2008, 299: 646-655. 10.1001/jama.299.6.646

    Article  CAS  PubMed  Google Scholar 

  44. Meade MO, Cook DJ, Guyatt GH, Slutsky AS, Arabi YM, Cooper DJ, Davies AR, Hand LE, Zhou Q, Thabane L, Austin P, Lapinsky S, Baxter A, Russell J, Skrobik Y, Ronco JJ, Stewart TE, Lung Open Ventilation Study Investigators: Ventilation strategy using low tidal volumes, recruitment maneuvers, and high positive end-expiratory pressure for acute lung injury and acute respiratory distress syndrome: a randomized controlled trial. JAMA 2008, 299: 637-645. 10.1001/jama.299.6.637

    Article  CAS  PubMed  Google Scholar 

Download references

Acknowledgements

We would like to express our gratitude to Mrs Michal Kedar for coordinating the data of this study. This research was funded in part by Deep Breeze Ltd.

Author information

Authors and Affiliations

Authors

Corresponding author

Correspondence to Shaul Lev.

Additional information

Competing interests

Research materials for the VRI research program at Rabin Medical Center (Petah Tikva, Israel) are funded partially by Deep Breeze Ltd. SL has consultant agreement that includes honoraria and stock options (no current monetary value) with Deep Breeze Ltd and he was sponsored by GE Healthcare, Deep Breeze's distributor worldwide, to give lectures in academic meetings. YAG is an employee of Deep Breeze Ltd. IK, DD, JC, MG, MS, and PS declare that they have no competing interests.

Authors' contributions

SL, YAG, IK, DD, MG, and MS participated in the design and coordination of the study and carried out the VRI recordings. SL and YAG worked on the data analysis. SL, YAG, JC, and PS drafted the manuscript. All authors edited and approved the final manuscript.

Authors’ original submitted files for images

Rights and permissions

Reprints and permissions

About this article

Cite this article

Lev, S., Glickman, Y.A., Kagan, I. et al. Changes in regional distribution of lung sounds as a function of positive end-expiratory pressure. Crit Care 13, R66 (2009). https://doi.org/10.1186/cc7871

Download citation

  • Received:

  • Accepted:

  • Published:

  • DOI: https://doi.org/10.1186/cc7871

Keywords